Immobilize bacteria

classic Classic list List threaded Threaded
22 messages Options
12
Mike Tighe Mike Tighe
Reply | Threaded
Open this post in threaded view
|

Immobilize bacteria

Does anybody have a good way to Immobilize bacteria without killing. We would like to do a Live/dead stain and our bugs like to float away!

Thanks!!
Mike
G. Esteban Fernandez G. Esteban Fernandez
Reply | Threaded
Open this post in threaded view
|

Re: Immobilize bacteria

I've used CyGel for that purpose before (no commercial interest).
http://www.biostatus.com/product/cygel/

-Esteban


On Thu, Oct 22, 2009 at 1:11 PM, Mike Tighe <[hidden email]> wrote:
> Does anybody have a good way to Immobilize bacteria without killing. We would like to do a Live/dead stain and our bugs like to float away!
>
> Thanks!!
> Mike
>

--
G. Esteban Fernandez, Ph.D.

Associate Director
Molecular Cytology Core Facility
University of Missouri
120 Bond Life Sciences Center
Columbia, MO  65211

http://www.biotech.missouri.edu/mcc

(573)882-4895
(573)884-9676 fax
O'Toole, P O'Toole, P
Reply | Threaded
Open this post in threaded view
|

Re: Immobilize bacteria

Likewise, we have used CyGel. You may need try various strengths to get the
best lifetime out of the bacteria, especially if they are motile. You will
also need to seal the edges with vacuum grease to stop it drying out, but
once in place.

Best

Peter

----- Original Message -----
From: "G. Esteban Fernandez" <[hidden email]>
To: <[hidden email]>
Sent: Thursday, October 22, 2009 7:17 PM
Subject: Re: Immobilize bacteria


> I've used CyGel for that purpose before (no commercial interest).
> http://www.biostatus.com/product/cygel/
>
> -Esteban
>
>
> On Thu, Oct 22, 2009 at 1:11 PM, Mike Tighe <[hidden email]>
> wrote:
>> Does anybody have a good way to Immobilize bacteria without killing. We
>> would like to do a Live/dead stain and our bugs like to float away!
>>
>> Thanks!!
>> Mike
>>
>
> --
> G. Esteban Fernandez, Ph.D.
>
> Associate Director
> Molecular Cytology Core Facility
> University of Missouri
> 120 Bond Life Sciences Center
> Columbia, MO  65211
>
> http://www.biotech.missouri.edu/mcc
>
> (573)882-4895
> (573)884-9676 fax
>
Colin Rickman-2 Colin Rickman-2
Reply | Threaded
Open this post in threaded view
|

Re: Immobilize bacteria

In reply to this post by Mike Tighe
I use CellTak from BD which works very well. Immobilises them well  
enough for long time cours records and they still grow fine.

Colin


On 22 Oct 2009, at 19:11, Mike Tighe <[hidden email]>  
wrote:

> Does anybody have a good way to Immobilize bacteria without killing.  
> We would like to do a Live/dead stain and our bugs like to float away!
>
> Thanks!!
> Mike
Heather A Owen Heather A Owen
Reply | Threaded
Open this post in threaded view
|

Re: Immobilize bacteria

In reply to this post by Mike Tighe
Hi Mike,

A trick taught to me my a microbiologist (Mark McBride) is to make a 0.7% agarose solution in water (microwave to dissolve), and while it is still hot use a Pasteur pipet to flow a layer onto a microscope slide to cover it.  I usully suspend the slides on wooden applicator sticks in case the agarose runs off, otherwise the solution wicks under the slide.  Once the agarose gels, apply a drop of culture on top, then coverslip it.  The pad of agarose needs to be larger than the coverslip you want to use.  The cells stay immobile, and actually stay alive for a long time.

Sincerely yours,
Heather Owen


Dr. Heather A. Owen
Director, Electron Microscope Laboratory
Department of Biological Sciences
University of Wisconsin - Milwaukee
Lapham Hall
3209 N. Maryland Avenue
Milwaukee, WI   53211
(414)229-6816

----- Original Message -----
From: "Mike Tighe" <[hidden email]>
To: [hidden email]
Sent: Thursday, October 22, 2009 1:11:00 PM GMT -06:00 US/Canada Central
Subject: Immobilize bacteria

Does anybody have a good way to Immobilize bacteria without killing. We would like to do a Live/dead stain and our bugs like to float away!

Thanks!!
Mike
rjpalmer rjpalmer
Reply | Threaded
Open this post in threaded view
|

Re: Immobilize bacteria

In reply to this post by Mike Tighe
The simplest approach is to just make a nice thin mount and seal the
edges with wax or nail polish.  Should work pretty well.  However,
for serious counting, you filter the bacteria onto a black filter,
apply oil to the filter, a coverslip on top, and then use oil
immersion optics to view.  Another thing you can try is to coat
slides with poly-lysine (I think you can buy pre-coated slides), add
a drop of the bugs, then coverslip the prep.  If you want to go
really low tech but get results, you can coat a slide with 0.5% agar
(tilt the slide and run the nearly cool agar down it), wait for it to
dry, then add a drop of bacterial suspension.  The agar swells, the
bacteria are immobilized, and you drop a coverglass on the top.

>Does anybody have a good way to Immobilize bacteria without killing.
>We would like to do a Live/dead stain and our bugs like to float
>away!
>
>Thanks!!
>Mike


--
Robert J. Palmer Jr., Ph.D.
Natl Inst Dental Craniofacial Res - Natl Insts Health
Oral Infection and Immunity Branch
Bldg 30, Room 310
30 Convent Drive
Bethesda MD 20892
ph 301-594-0025
fax 301-402-0396
B. Prabhakar Pandian B. Prabhakar Pandian
Reply | Threaded
Open this post in threaded view
|

Cell Media Preparation Debate

In reply to this post by O'Toole, P
Hello,
             In our laboratory there has been recent debates on the cell
media preparation. The point of contention is the presence of serum in
media.
When a media mentions supplemented with 10% serum, which of the
following is correct for a 500ml media bottle.

1. Add 50ml of serum to the media bottle (total volume is 550ml)

2. Remove 50ml of media and then add 50ml of serum (total volume is 500ml).

Any information on the right usage will be great.

Thanks,

-Bala
Michael Herron Michael Herron
Reply | Threaded
Open this post in threaded view
|

Re: Immobilize bacteria

In reply to this post by O'Toole, P
Is there a U.S, vendor for CyGel?


On Oct 22, 2009, at 1:31 PM, Peter O'Toole wrote:

> Likewise, we have used CyGel. You may need try various strengths to  
> get the best lifetime out of the bacteria, especially if they are  
> motile. You will also need to seal the edges with vacuum grease to  
> stop it drying out, but once in place.
>
> Best
>
> Peter
>
> ----- Original Message ----- From: "G. Esteban Fernandez" <[hidden email]
> >
> To: <[hidden email]>
> Sent: Thursday, October 22, 2009 7:17 PM
> Subject: Re: Immobilize bacteria
>
>
>> I've used CyGel for that purpose before (no commercial interest).
>> http://www.biostatus.com/product/cygel/
>>
>> -Esteban
>>
>>
>> On Thu, Oct 22, 2009 at 1:11 PM, Mike Tighe <[hidden email]
>> > wrote:
>>> Does anybody have a good way to Immobilize bacteria without  
>>> killing. We would like to do a Live/dead stain and our bugs like  
>>> to float away!
>>>
>>> Thanks!!
>>> Mike
>>>
>>
>> --
>> G. Esteban Fernandez, Ph.D.
>>
>> Associate Director
>> Molecular Cytology Core Facility
>> University of Missouri
>> 120 Bond Life Sciences Center
>> Columbia, MO  65211
>>
>> http://www.biotech.missouri.edu/mcc
>>
>> (573)882-4895
>> (573)884-9676 fax

Michael J. Herron,  U of MN, Dept. of Entomology
   [hidden email]
      612-624-3688 (office) 612-625-5299 (FAX)
Withers, Helen Withers, Helen
Reply | Threaded
Open this post in threaded view
|

Re: Immobilize bacteria

In reply to this post by Mike Tighe
In the past we have used thin layers of agar or agarose on the surface of the slide to immobilise either pre-stain bacteria or to actually stain bacteria by incorporating the stain into the agar. This effectively prevents floating if you lightly dry the surface of the agar once the sample has been added prior to putting on a coverslip. This works very well for both live/dead staining and for time-lapse growth studies using DAPI or membrane stained bacteria so they remain viable for long periods (assuming that they are in a humidified container!).

Helen

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mike Tighe
Sent: Friday, 23 October 2009 7:11 a.m.
To: [hidden email]
Subject: Immobilize bacteria

Does anybody have a good way to Immobilize bacteria without killing. We would like to do a Live/dead stain and our bugs like to float away!

Thanks!!
Mike
=======================================================================
Attention: The information contained in this message and/or attachments
from AgResearch Limited is intended only for the persons or entities
to which it is addressed and may contain confidential and/or privileged
material. Any review, retransmission, dissemination or other use of, or
taking of any action in reliance upon, this information by persons or
entities other than the intended recipients is prohibited by AgResearch
Limited. If you have received this message in error, please notify the
sender immediately.
=======================================================================
Mike Tighe Mike Tighe
Reply | Threaded
Open this post in threaded view
|

Re: Cell Media Preparation Debate

In reply to this post by B. Prabhakar Pandian
If total volume is 550mls (500ml media and 50mls serum you will have  9.1%  serum. Not 10% but probably enough to keep most cells happy

Mike
 
>>> "B. Prabhakar Pandian" <[hidden email]> 10/22/09 3:10 PM >>>
Hello,
             In our laboratory there has been recent debates on the cell
media preparation. The point of contention is the presence of serum in
media.
When a media mentions supplemented with 10% serum, which of the
following is correct for a 500ml media bottle.

1. Add 50ml of serum to the media bottle (total volume is 550ml)

2. Remove 50ml of media and then add 50ml of serum (total volume is 500ml).

Any information on the right usage will be great.

Thanks,

- Bala
Henn, Alicia Henn, Alicia
Reply | Threaded
Open this post in threaded view
|

Re: Cell Media Preparation Debate

In reply to this post by B. Prabhakar Pandian
We add 55ml of serum to a 500ml bottle of medium to yield 555ml of 10% serum. No 50ml tubes of extra medium to store and still get 10% final.

Alicia D. Henn, Ph.D.
Department of Nephrology
University of Rochester Medical Center
Rochester, NY 14624
(585) 276-4897

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of B. Prabhakar Pandian
Sent: Thursday, October 22, 2009 3:11 PM
To: [hidden email]
Subject: Cell Media Preparation Debate

Hello,
             In our laboratory there has been recent debates on the cell
media preparation. The point of contention is the presence of serum in
media.
When a media mentions supplemented with 10% serum, which of the
following is correct for a 500ml media bottle.

1. Add 50ml of serum to the media bottle (total volume is 550ml)

2. Remove 50ml of media and then add 50ml of serum (total volume is 500ml).

Any information on the right usage will be great.

Thanks,

-Bala
Nobby Clark Nobby Clark
Reply | Threaded
Open this post in threaded view
|

Re: Cell Media Preparation Debate

In reply to this post by Mike Tighe
Hi,
Does anybody no of anyone who is looking for a very nice confocial? we have
lost funding and must get rid of our Zeiss LSM 510NLO ready. We must find a
home for this soon
Thanks



On Thu, 22 Oct 2009 16:40:17 -0400, Mike Tighe
<[hidden email]>
wrote:
> If total volume is 550mls (500ml media and 50mls serum you will have
9.1%

> serum. Not 10% but probably enough to keep most cells happy
>
> Mike
>  
>>>> "B. Prabhakar Pandian" <[hidden email]> 10/22/09 3:10 PM >>>
> Hello,
>              In our laboratory there has been recent debates on the cell
> media preparation. The point of contention is the presence of serum in
> media.
> When a media mentions supplemented with 10% serum, which of the
> following is correct for a 500ml media bottle.
>
> 1. Add 50ml of serum to the media bottle (total volume is 550ml)
>
> 2. Remove 50ml of media and then add 50ml of serum (total volume is
500ml).
>
> Any information on the right usage will be great.
>
> Thanks,
>
> - Bala
B. Prabhakar Pandian B. Prabhakar Pandian
Reply | Threaded
Open this post in threaded view
|

Re: Cell Media Preparation Debate

In reply to this post by Mike Tighe
Hello Mike,
                       I know that from the mathematical stand point
what you are saying is correct, but the contention is more from what 10%
actually means. Is it to add 10% of the original volume of
serum (option # 1) to the media or to make sure that the final
concentration is 10% (option # 2).

Thanks,

Mike Tighe wrote:

> If total volume is 550mls (500ml media and 50mls serum you will have  9.1%  serum. Not 10% but probably enough to keep most cells happy
>
> Mike
>  
>  
>>>> "B. Prabhakar Pandian" <[hidden email]> 10/22/09 3:10 PM >>>
>>>>        
> Hello,
>              In our laboratory there has been recent debates on the cell
> media preparation. The point of contention is the presence of serum in
> media.
> When a media mentions supplemented with 10% serum, which of the
> following is correct for a 500ml media bottle.
>
> 1. Add 50ml of serum to the media bottle (total volume is 550ml)
>
> 2. Remove 50ml of media and then add 50ml of serum (total volume is 500ml).
>
> Any information on the right usage will be great.
>
> Thanks,
>
> - Bala
>
>
>  


--
Mike Tighe Mike Tighe
Reply | Threaded
Open this post in threaded view
|

Re: Cell Media Preparation Debate

I think that option #2 is the intent.
 
>>> "B. Prabhakar Pandian" <[hidden email]> 10/22/09 4:54 PM >>>
Hello Mike,
                       I know that from the mathematical stand point
what you are saying is correct, but the contention is more from what 10%
actually means. Is it to add 10% of the original volume of
serum (option # 1) to the media or to make sure that the final
concentration is 10% (option # 2).

Thanks,

Mike Tighe wrote:

> If total volume is 550mls (500ml media and 50mls serum you will have  9.1%  serum. Not 10% but probably enough to keep most cells happy
>
> Mike
>  
>  
>>>> "B. Prabhakar Pandian" <[hidden email]> 10/22/09 3:10 PM >>>
>>>>        
> Hello,
>              In our laboratory there has been recent debates on the cell
> media preparation. The point of contention is the presence of serum in
> media.
> When a media mentions supplemented with 10% serum, which of the
> following is correct for a 500ml media bottle.
>
> 1. Add 50ml of serum to the media bottle (total volume is 550ml)
>
> 2. Remove 50ml of media and then add 50ml of serum (total volume is 500ml).
>
> Any information on the right usage will be great.
>
> Thanks,
>
> -  Bala
>
>
>  


--  
Phillips, Thomas E. Phillips, Thomas E.
Reply | Threaded
Open this post in threaded view
|

Re: Cell Media Preparation Debate

The point is moot. 10% anything in a scientific formulation generally implies it was a somewhat arbitrarily chosen concentration that works well for many things. If you really want to know the correct concentration, grow your cells in multiple concentrations and see which is best. It is foolish to agonize over an arbitrary value. If it isn't worth optimizing, it isn't worth worrying about whether you are using 9.1% or 10%. You can publish the precise value (e.g., 50 mls of FBS in 500 mls DME) so that your work can be replicated if you want to stop the ambiguity of this type of concentration. My experience is that many cells grow as well in 5% FBS as in 10%.  


Thomas E. Phillips, Ph.D
Professor of Biological Sciences
Director, Molecular Cytology Core
2 Tucker Hall
University of Missouri
Columbia, MO 65211-7400
573-882-4712 (office)
573-882-0123 (fax)
[hidden email]

http://www.biology.missouri.edu/faculty/phillips.html
http://www.biotech.missouri.edu/mcc/


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mike Tighe
Sent: Thursday, October 22, 2009 4:17 PM
To: [hidden email]
Subject: Re: Cell Media Preparation Debate

I think that option #2 is the intent.
 
>>> "B. Prabhakar Pandian" <[hidden email]> 10/22/09 4:54 PM >>>
Hello Mike,
                       I know that from the mathematical stand point
what you are saying is correct, but the contention is more from what 10%
actually means. Is it to add 10% of the original volume of
serum (option # 1) to the media or to make sure that the final
concentration is 10% (option # 2).

Thanks,

Mike Tighe wrote:

> If total volume is 550mls (500ml media and 50mls serum you will have  9.1%  serum. Not 10% but probably enough to keep most cells happy
>
> Mike
>  
>  
>>>> "B. Prabhakar Pandian" <[hidden email]> 10/22/09 3:10 PM >>>
>>>>        
> Hello,
>              In our laboratory there has been recent debates on the cell
> media preparation. The point of contention is the presence of serum in
> media.
> When a media mentions supplemented with 10% serum, which of the
> following is correct for a 500ml media bottle.
>
> 1. Add 50ml of serum to the media bottle (total volume is 550ml)
>
> 2. Remove 50ml of media and then add 50ml of serum (total volume is 500ml).
>
> Any information on the right usage will be great.
>
> Thanks,
>
> -  Bala
>
>
>  


--  
Boswell, Carl A - (cboswell) Boswell, Carl A - (cboswell)
Reply | Threaded
Open this post in threaded view
|

Re: Cell Media Preparation Debate

In reply to this post by B. Prabhakar Pandian
Option #2 is always the intent in this context.  The goal is the FINAL
concentration of whatever % you want.

C

Carl A. Boswell, Ph.D.
Molecular and Cellular Biology
University of Arizona
520-954-7053
FAX 520-621-3709
----- Original Message -----
From: "B. Prabhakar Pandian" <[hidden email]>
To: <[hidden email]>
Sent: Thursday, October 22, 2009 1:54 PM
Subject: Re: Cell Media Preparation Debate


> Hello Mike,
>                       I know that from the mathematical stand point what
> you are saying is correct, but the contention is more from what 10%
> actually means. Is it to add 10% of the original volume of
> serum (option # 1) to the media or to make sure that the final
> concentration is 10% (option # 2).
>
> Thanks,
>
> Mike Tighe wrote:
>> If total volume is 550mls (500ml media and 50mls serum you will have
>> 9.1%  serum. Not 10% but probably enough to keep most cells happy
>>
>> Mike
>>
>>>>> "B. Prabhakar Pandian" <[hidden email]> 10/22/09 3:10 PM >>>
>> Hello,
>>              In our laboratory there has been recent debates on the cell
>> media preparation. The point of contention is the presence of serum in
>> media.
>> When a media mentions supplemented with 10% serum, which of the following
>> is correct for a 500ml media bottle.
>>
>> 1. Add 50ml of serum to the media bottle (total volume is 550ml)
>>
>> 2. Remove 50ml of media and then add 50ml of serum (total volume is
>> 500ml).
>>
>> Any information on the right usage will be great.
>>
>> Thanks,
>>
>> - Bala
>>
>>
>>
>
>
> --
>
Martin Wessendorf-2 Martin Wessendorf-2
Reply | Threaded
Open this post in threaded view
|

Re: Cell Media Preparation Debate

I wonder to what extent the present confusion is due to Kodak.  When
making developer, fixer, etc., a "1:4" dilution always meant one part
ADDED TO four parts.

Martin

Carl Boswell wrote:

> Option #2 is always the intent in this context.  The goal is the FINAL
> concentration of whatever % you want.
>
> C
>
> Carl A. Boswell, Ph.D.
> Molecular and Cellular Biology
> University of Arizona
> 520-954-7053
> FAX 520-621-3709
> ----- Original Message ----- From: "B. Prabhakar Pandian" <[hidden email]>
> To: <[hidden email]>
> Sent: Thursday, October 22, 2009 1:54 PM
> Subject: Re: Cell Media Preparation Debate
>
>
>> Hello Mike,
>>                       I know that from the mathematical stand point
>> what you are saying is correct, but the contention is more from what
>> 10% actually means. Is it to add 10% of the original volume of
>> serum (option # 1) to the media or to make sure that the final
>> concentration is 10% (option # 2).
>>
>> Thanks,
>>
>> Mike Tighe wrote:
>>> If total volume is 550mls (500ml media and 50mls serum you will have
>>> 9.1%  serum. Not 10% but probably enough to keep most cells happy
>>>
>>> Mike
>>>
>>>>>> "B. Prabhakar Pandian" <[hidden email]> 10/22/09 3:10 PM >>>
>>> Hello,
>>>              In our laboratory there has been recent debates on the
>>> cell media preparation. The point of contention is the presence of
>>> serum in media.
>>> When a media mentions supplemented with 10% serum, which of the
>>> following is correct for a 500ml media bottle.
>>>
>>> 1. Add 50ml of serum to the media bottle (total volume is 550ml)
>>>
>>> 2. Remove 50ml of media and then add 50ml of serum (total volume is
>>> 500ml).
>>>
>>> Any information on the right usage will be great.
>>>
>>> Thanks,
>>>
>>> - Bala
>>>
>>>
>>>
>>
>>
>> --
>>

--
Martin Wessendorf, Ph.D.                   office: (612) 626-0145
Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991
University of Minnesota             Preferred FAX: (612) 624-8118
6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009
Minneapolis, MN  55455                    e-mail: [hidden email]
Bruno Afonso Bruno Afonso
Reply | Threaded
Open this post in threaded view
|

Re: Immobilize bacteria

In reply to this post by Withers, Helen
To get an humidified environment just use a glass bottom mattek dish with some water droplets on the outer edges. Cells can grow on an agarose pad for days. 

b


On Thu, Oct 22, 2009 at 16:11, Withers, Helen <[hidden email]> wrote:
In the past we have used thin layers of agar or agarose on the surface of the slide to immobilise either pre-stain bacteria or to actually stain bacteria by incorporating the stain into the agar. This effectively prevents floating if you lightly dry the surface of the agar once the sample has been added prior to putting on a coverslip. This works very well for both live/dead staining and for time-lapse growth studies using DAPI or membrane stained bacteria so they remain viable for long periods (assuming that they are in a humidified container!).

Helen

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mike Tighe
Sent: Friday, 23 October 2009 7:11 a.m.
To: [hidden email]
Subject: Immobilize bacteria

Does anybody have a good way to Immobilize bacteria without killing. We would like to do a Live/dead stain and our bugs like to float away!

Thanks!!
Mike
=======================================================================
Attention: The information contained in this message and/or attachments
from AgResearch Limited is intended only for the persons or entities
to which it is addressed and may contain confidential and/or privileged
material. Any review, retransmission, dissemination or other use of, or
taking of any action in reliance upon, this information by persons or
entities other than the intended recipients is prohibited by AgResearch
Limited. If you have received this message in error, please notify the
sender immediately.
=======================================================================


Jennifer Clarke Jennifer Clarke
Reply | Threaded
Open this post in threaded view
|

Re: Immobilize bacteria

In reply to this post by Heather A Owen
Hi Mike

I can vouch for the agarose method and have taken good confocal images this way
of live immobilised bacterial cells.  Ensure you prepare the agarose slides on
a level bench just before use as specified by Heather below

Kind regards
Jen

Quoting Heather A Owen <[hidden email]>:

> Hi Mike,
>
> A trick taught to me my a microbiologist (Mark McBride) is to make a 0.7%
> agarose solution in water (microwave to dissolve), and while it is still hot
> use a Pasteur pipet to flow a layer onto a microscope slide to cover it.  I
> usully suspend the slides on wooden applicator sticks in case the agarose
> runs off, otherwise the solution wicks under the slide.  Once the agarose
> gels, apply a drop of culture on top, then coverslip it.  The pad of agarose
> needs to be larger than the coverslip you want to use.  The cells stay
> immobile, and actually stay alive for a long time.
>
> Sincerely yours,
> Heather Owen
>
>
> Dr. Heather A. Owen
> Director, Electron Microscope Laboratory
> Department of Biological Sciences
> University of Wisconsin - Milwaukee
> Lapham Hall
> 3209 N. Maryland Avenue
> Milwaukee, WI   53211
> (414)229-6816
>
> ----- Original Message -----
> From: "Mike Tighe" <[hidden email]>
> To: [hidden email]
> Sent: Thursday, October 22, 2009 1:11:00 PM GMT -06:00 US/Canada Central
> Subject: Immobilize bacteria
>
> Does anybody have a good way to Immobilize bacteria without killing. We would
> like to do a Live/dead stain and our bugs like to float away!
>
> Thanks!!
> Mike
>


--
Jennifer Clarke BSc (Hons) PhD
Research Associate, Anatomy and Histology
Centre for Neuroscience, School of Medicine
&
Facility Manager, Optical Microscopy Suite, Flinders Microscopy
(available for training and assistance on Mondays only)

Flinders University
GPO Box 2100, Adelaide 5001
Phone: 61 8 8204 6454 / 61 8 8204 4205
Email: [hidden email]
Roy Edward Roy Edward
Reply | Threaded
Open this post in threaded view
|

Re: Immobilize bacteria

In reply to this post by Mike Tighe
Dear Michael

You have two options for ordering CyGEL (Product code: CY10500) (& DRAQ5,
etc):

1) our website www.biostatus.com: go to our "purchase" page, register and
then if USA is used the prices come up in USD.  You can order by credit
card or conventional purchase order system using the guidance notes.  We
deliver anywhere worldwide within 48 hours.

2) Axxora.com: CyGEL is distributed by Axxora.  Check the price before
ordering cf. our published web prices.

You may be interested to know that CyGEL was cited in a new methods paper
published online last week for live parasite imaging from the laboratory
of Professor Deborah Smith, University of York, UK:

Validation of a New Method for Immobilising Kinetoplastid Parasites for
Live Cell Imaging. Helen P. Price, et al. Molecular and Biological
Parasitology

http://www.sciencedirect.com/science?_ob=ArticleURL&_udi=B6T29-4XCYJ8M-1&_user=4306206&_rdoc=1&_fmt=&_orig=search&_sort=d&_docanchor=&view=c&_acct=C000010338&_version=1&_urlVersion=0&_userid=4306206&md5=c8b8f4f1cf55421b0f7de62d46fd3af6

Don't hesitate to contact me if you have any questions regarding CyGEL or
CyGEL Sustain, particularly regarding the creation of wells on slides, or
imaging of cells, C elegans, Danio, etc.

Best regards
Roy Edward

Biostatus Limited
Tel: +44 (0)1509 558 163
Fax: +44 (0)1509 651 061
www.biostatus.com
Company No: 3079239.
Registered in England and Wales.
Registered office: Biostatus Ltd, 56 Charnwood Road, Shepshed,
Leicestershire, LE12 9NP, UK.


-----Original Message-----
From: Michael Herron [mailto:[hidden email]]
Sent: 22 October 2009 20:54
Subject: Re: Immobilize bacteria

Is there a U.S, vendor for CyGel?


On Oct 22, 2009, at 1:31 PM, Peter O'Toole wrote:

> Likewise, we have used CyGel. You may need try various strengths to
> get the best lifetime out of the bacteria, especially if they are
> motile. You will also need to seal the edges with vacuum grease to
> stop it drying out, but once in place.
>
> Best
>
> Peter
>
> ----- Original Message ----- From: "G. Esteban Fernandez"
> <[hidden email]
> >
> To: <[hidden email]>
> Sent: Thursday, October 22, 2009 7:17 PM
> Subject: Re: Immobilize bacteria
>
>
>> I've used CyGel for that purpose before (no commercial interest).
>> http://www.biostatus.com/product/cygel/
>>
>> -Esteban
>>
>>
>> On Thu, Oct 22, 2009 at 1:11 PM, Mike Tighe
>> <[hidden email]
>> > wrote:
>>> Does anybody have a good way to Immobilize bacteria without killing.
>>> We would like to do a Live/dead stain and our bugs like to float
>>> away!
>>>
>>> Thanks!!
>>> Mike
>>>
>>
>> --
>> G. Esteban Fernandez, Ph.D.
>>
>> Associate Director
>> Molecular Cytology Core Facility
>> University of Missouri
>> 120 Bond Life Sciences Center
>> Columbia, MO  65211
>>
>> http://www.biotech.missouri.edu/mcc
>>
>> (573)882-4895
>> (573)884-9676 fax

Michael J. Herron,  U of MN, Dept. of Entomology
   [hidden email]
      612-624-3688 (office) 612-625-5299 (FAX)
12