Jeffrey Carmichael |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** COMMERCIAL POST In a recent conversation thread about blue fluorochromes, George McNamara mentioned alternative labeling methods such as biotin/streptavidin and TSA-hapten. This reminded me of a customer we've been working with who has developed what they're calling a "next generation" hapten (not TSA) labeling technology with very high specificity, affinity and amplification. This in turn means that the primary antibody species origin doesn't matter, and 4 mouse monoclonal antibodies may be used together on the same tissue sample, including without multiple labeling steps. The "modified" haptens are referred to as peptide bar-codes which apparently allow for significant multiplexing: https://cellidx.com/technology https://event.webcasts.com/viewer/event.jsp?ei=1187871&tp_key=9db65f5892 Chroma Technology has no commercial interest in these products per se, other than the filter sets that may be recommended for use with them. Jeff *Jeff Carmichael* *Product & Technical Marketing Manager* *[hidden email] <[hidden email]> | 802-428-2528* * <[hidden email]>* -- <https://www.chroma.com/>CHROMA TECHNOLOGY CORP® *an employee owned company* 10 Imtec Lane, Bellows Falls, Vermont 05101 USA 800-824-7662 | FAX: 802-428-2525 www.chroma.com <https://www.chroma.com/> | [hidden email] <mailto:[hidden email]> |
|
George McNamara |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** As much as I am a fan of TSA (and look forward to the methyl-luminol alternative), more multiplexing options: https://www.ncbi.nlm.nih.gov/pubmed/29263082 Pleiner et al 2018 JCB -A toolbox of anti-mouse and anti-rabbit IgG secondary nanobodies. https://www.ncbi.nlm.nih.gov/pubmed/29444803 Commentary The commercial arm is www.nano-tag.com In many ways, nanobodies are an update of the "Pretty Poly" method of Morris and Stanley 2003 https://www.ncbi.nlm.nih.gov/pubmed/?term=pretty+poly+stanley see also Molecular Probes Zenon product line. These are not amplification. See also Brilliants and SuperBrights: ideally the fluorescence microscopy community could be using the direct labeled fluorescent antibodies routinely used in flow cytometers and FACE sorter. I mentioned to some visitors today that BD's X50 is named for their intent (and look to be on target to achieve) 50plex fluorescence flow cytometry, with several directional cell sorting (I'm spacing out on the numbers, maybe 6 way sort now, 10 way in ~3 years?). I'm also a fan of CyTOF, imaging CyTOF, MIBI-ToF, etcTOF. enjoy, George p.s. Disclosure: I came across nano-tag.com at their exhibit at Society for Neuroscience 2017, who turned me on to the Pleiner article. They were nice enough to send several free samples, which our users found worked fine on our new Leica SP8 confocal microscope (should also work fine on our Olympus FV3000RS, but that only was installed last week). On 8/23/2018 4:57 PM, Jeff Carmichael wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > COMMERCIAL POST > > In a recent conversation thread about blue fluorochromes, George McNamara > mentioned alternative labeling methods such as biotin/streptavidin and > TSA-hapten. > > This reminded me of a customer we've been working with who has developed > what they're calling a "next generation" hapten (not TSA) labeling > technology with very high specificity, affinity and amplification. This in > turn means that the primary antibody species origin doesn't matter, and 4 > mouse monoclonal antibodies may be used together on the same tissue sample, > including without multiple labeling steps. The "modified" haptens are > referred to as peptide bar-codes which apparently allow for significant > multiplexing: > > https://cellidx.com/technology > https://event.webcasts.com/viewer/event.jsp?ei=1187871&tp_key=9db65f5892 > > Chroma Technology has no commercial interest in these products per se, > other than the filter sets that may be recommended for use with them. > > Jeff > > *Jeff Carmichael* > *Product & Technical Marketing Manager* > > *[hidden email] <[hidden email]> | 802-428-2528* > > * <[hidden email]>* > -- George McNamara, PhD Baltimore, MD 21231 [hidden email] https://www.linkedin.com/in/georgemcnamara https://works.bepress.com/gmcnamara/75 (may need to use Microsoft Edge or Firefox, rather than Google Chrome) http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650 http://confocal.jhu.edu July 2017 Current Protocols article, open access: UNIT 4.4 Microscopy and Image Analysis http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/abstract supporting materials direct link is http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/full#hg0404-sec-0023 figures at http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/figures |
|
Sripad Ram-2 |
|
|
In reply to this post by Jeffrey Carmichael
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Jeff, My understanding is that cellIdx technology can support upto a 4plex (i.e. 4 abs). I dont recall seeing side by side comparison of their technology with TSA, but what i hear is that the signal intensity is good enough to detect Ab labeling even in notoriously autofluorescent samples (lung cancer tissue). Sripad On Thu, Aug 23, 2018, 2:02 PM Jeff Carmichael <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > COMMERCIAL POST > > In a recent conversation thread about blue fluorochromes, George McNamara > mentioned alternative labeling methods such as biotin/streptavidin and > TSA-hapten. > > This reminded me of a customer we've been working with who has developed > what they're calling a "next generation" hapten (not TSA) labeling > technology with very high specificity, affinity and amplification. This in > turn means that the primary antibody species origin doesn't matter, and 4 > mouse monoclonal antibodies may be used together on the same tissue sample, > including without multiple labeling steps. The "modified" haptens are > referred to as peptide bar-codes which apparently allow for significant > multiplexing: > > https://cellidx.com/technology > https://event.webcasts.com/viewer/event.jsp?ei=1187871&tp_key=9db65f5892 > > Chroma Technology has no commercial interest in these products per se, > other than the filter sets that may be recommended for use with them. > > Jeff > > *Jeff Carmichael* > *Product & Technical Marketing Manager* > > *[hidden email] <[hidden email]> | 802-428-2528* > > * <[hidden email]>* > > -- > <https://www.chroma.com/>CHROMA TECHNOLOGY CORP® > *an employee owned > company* > 10 Imtec Lane, Bellows Falls, Vermont 05101 USA > 800-824-7662 | > FAX: 802-428-2525 > www.chroma.com <https://www.chroma.com/> | > [hidden email] <mailto:[hidden email]> > |
|
Martin Wessendorf-2 |
|
|
In reply to this post by George McNamara
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I'm going to take issue with George on one point: I don't think that directly labeled primaries are necessarily the best way to go. The beauty of indirect immunofluorescence is that you can combine a given primary antibody with any number of different fluorophores (in the form of fluorophore-conjugated secondary antibodies). Once you've characterized that primary antibody, you can be reasonably sure that it will label the same molecule no matter what color secondary antibody it's paired with (...assuming that the secondary works well enough to detect any labeling!). In contrast, each different fluorophore-conjugate of a directly conjugated primary antibody, is a chemically different reagent that may have different binding properties and will need to be re-characterized for specificity. --I.e., you would be setting yourself up for a lot of extra work if you use directly conjugated primaries. (If you've already characterized the specific reagents you'll be using, then this isn't an issue.) My two-cents worth! Martin Wessendorf On 8/23/2018 8:43 PM, George McNamara wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > As much as I am a fan of TSA (and look forward to the methyl-luminol > alternative), more multiplexing options: > > https://www.ncbi.nlm.nih.gov/pubmed/29263082 > > Pleiner et al 2018 JCB -A toolbox of anti-mouse and anti-rabbit IgG > secondary nanobodies. > > https://www.ncbi.nlm.nih.gov/pubmed/29444803 > > Commentary > > The commercial arm is www.nano-tag.com > > In many ways, nanobodies are an update of the "Pretty Poly" method of > Morris and Stanley 2003 > https://www.ncbi.nlm.nih.gov/pubmed/?term=pretty+poly+stanley > > see also Molecular Probes Zenon product line. > > These are not amplification. > > See also Brilliants and SuperBrights: ideally the fluorescence > microscopy community could be using the direct labeled fluorescent > antibodies routinely used in flow cytometers and FACE sorter. I > mentioned to some visitors today that BD's X50 is named for their > intent (and look to be on target to achieve) 50plex fluorescence flow > cytometry, with several directional cell sorting (I'm spacing out on > the numbers, maybe 6 way sort now, 10 way in ~3 years?). I'm also a > fan of CyTOF, imaging CyTOF, MIBI-ToF, etcTOF. > > enjoy, > George > p.s. Disclosure: I came across nano-tag.com at their exhibit at > Society for Neuroscience 2017, who turned me on to the Pleiner > article. They were nice enough to send several free samples, which our > users found worked fine on our new Leica SP8 confocal microscope > (should also work fine on our Olympus FV3000RS, but that only was > installed last week). > > On 8/23/2018 4:57 PM, Jeff Carmichael wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your >> posting. >> ***** >> >> COMMERCIAL POST >> >> In a recent conversation thread about blue fluorochromes, George >> McNamara >> mentioned alternative labeling methods such as biotin/streptavidin and >> TSA-hapten. >> >> This reminded me of a customer we've been working with who has developed >> what they're calling a "next generation" hapten (not TSA) labeling >> technology with very high specificity, affinity and amplification. >> This in >> turn means that the primary antibody species origin doesn't matter, >> and 4 >> mouse monoclonal antibodies may be used together on the same tissue >> sample, >> including without multiple labeling steps. The "modified" haptens are >> referred to as peptide bar-codes which apparently allow for significant >> multiplexing: >> >> https://cellidx.com/technology >> https://event.webcasts.com/viewer/event.jsp?ei=1187871&tp_key=9db65f5892 >> >> Chroma Technology has no commercial interest in these products per se, >> other than the filter sets that may be recommended for use with them. >> >> Jeff >> >> *Jeff Carmichael* >> *Product & Technical Marketing Manager* >> >> *[hidden email] <[hidden email]> | 802-428-2528* >> >> * <[hidden email]>* >> > |
|
George McNamara |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Martin, I chickened out on including (Komodo) dragon IgG in the (thought experiment, not 'reduced to practice') table from our immunofluorescence microscopy chapter (2011, book 1 chp 15 of https://www.cshlpress.com/default.tpl?action=full&--eqskudatarq=872&typ=ps&newtitle=Imaging%3A%20A%20Laboratory%20Manual ... recycled from an earlier imaging book) Tubulin Mouse IgG1 Goat–anti-mouse IgG1 Alexa Fluor 488 Green Actin Guinea pig IgG Goat–anti-guinea pig-IgG Lucifer Yellow Yellow Vimentin Rat mAb IgM Goat–anti-rat-IgM Cy3B Orange Cytokeratin Mouse IgG2a Goat–anti-mouse IgG2a DyLight 594 (or Texas Red) Red Neurofilament Chicken IgY Goat–anti-chicken IgY HiLyte Fluor 647 (or Cy5) Far red Nestin Rabbit IgG Goat–anti-rabbit IgG Atto 680 Infrared CD31 Donkey IgG Goat–anti-donkey IgG IRDye800 Infrared Quite unwieldy! for comparison, clinical flow cytometry is now 'routinely' done with a lot more than 7plex direct labeled antibodies. As for tissue section autofluorescence vs direct labeled antibody brightness: there are now many ways to decrease autofluorescence before imaging (I seem to be getting ads from Biotium once a week about their stuff, for example ... various authors have published using UV transilluminator to good and/or poor effect). There is also potential for new (likely expensive) fluorescence microscopes, such as Leica SP8 FALCON (disclosure: our image core hosts FALCON workshop week of Sept 10, 2018 ... hmmm: maybe I should measure BV421 lifetime), or other Fast FLIM instruments (whether upgrades or new). For examples, if NADH, collagen, and elastin autofluorescence lifetimes are 1 nanosecond, and organic dyes of similar spectra are each ~4 ns, could be 6plex just in blues and greens, plenty of spectra for more. enjoy, George On 8/23/2018 10:33 PM, Martin Wessendorf wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I'm going to take issue with George on one point: I don't think that > directly labeled primaries are necessarily the best way to go. The > beauty of indirect immunofluorescence is that you can combine a given > primary antibody with any number of different fluorophores (in the > form of fluorophore-conjugated secondary antibodies). Once you've > characterized that primary antibody, you can be reasonably sure that > it will label the same molecule no matter what color secondary > antibody it's paired with (...assuming that the secondary works well > enough to detect any labeling!). In contrast, each different > fluorophore-conjugate of a directly conjugated primary antibody, is a > chemically different reagent that may have different binding > properties and will need to be re-characterized for specificity. > --I.e., you would be setting yourself up for a lot of extra work if > you use directly conjugated primaries. (If you've already > characterized the specific reagents you'll be using, then this isn't > an issue.) > > My two-cents worth! > > Martin Wessendorf > > > > On 8/23/2018 8:43 PM, George McNamara wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your >> posting. >> ***** >> >> As much as I am a fan of TSA (and look forward to the methyl-luminol >> alternative), more multiplexing options: >> >> https://www.ncbi.nlm.nih.gov/pubmed/29263082 >> >> Pleiner et al 2018 JCB -A toolbox of anti-mouse and anti-rabbit IgG >> secondary nanobodies. >> >> https://www.ncbi.nlm.nih.gov/pubmed/29444803 >> >> Commentary >> >> The commercial arm is www.nano-tag.com >> >> In many ways, nanobodies are an update of the "Pretty Poly" method of >> Morris and Stanley 2003 >> https://www.ncbi.nlm.nih.gov/pubmed/?term=pretty+poly+stanley >> >> see also Molecular Probes Zenon product line. >> >> These are not amplification. >> >> See also Brilliants and SuperBrights: ideally the fluorescence >> microscopy community could be using the direct labeled fluorescent >> antibodies routinely used in flow cytometers and FACE sorter. I >> mentioned to some visitors today that BD's X50 is named for their >> intent (and look to be on target to achieve) 50plex fluorescence flow >> cytometry, with several directional cell sorting (I'm spacing out on >> the numbers, maybe 6 way sort now, 10 way in ~3 years?). I'm also a >> fan of CyTOF, imaging CyTOF, MIBI-ToF, etcTOF. >> >> enjoy, >> George >> p.s. Disclosure: I came across nano-tag.com at their exhibit at >> Society for Neuroscience 2017, who turned me on to the Pleiner >> article. They were nice enough to send several free samples, which >> our users found worked fine on our new Leica SP8 confocal microscope >> (should also work fine on our Olympus FV3000RS, but that only was >> installed last week). >> >> On 8/23/2018 4:57 PM, Jeff Carmichael wrote: >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> Post images on http://www.imgur.com and include the link in your >>> posting. >>> ***** >>> >>> COMMERCIAL POST >>> >>> In a recent conversation thread about blue fluorochromes, George >>> McNamara >>> mentioned alternative labeling methods such as biotin/streptavidin and >>> TSA-hapten. >>> >>> This reminded me of a customer we've been working with who has >>> developed >>> what they're calling a "next generation" hapten (not TSA) labeling >>> technology with very high specificity, affinity and amplification. >>> This in >>> turn means that the primary antibody species origin doesn't matter, >>> and 4 >>> mouse monoclonal antibodies may be used together on the same tissue >>> sample, >>> including without multiple labeling steps. The "modified" haptens are >>> referred to as peptide bar-codes which apparently allow for significant >>> multiplexing: >>> >>> https://cellidx.com/technology >>> https://event.webcasts.com/viewer/event.jsp?ei=1187871&tp_key=9db65f5892 >>> >>> >>> Chroma Technology has no commercial interest in these products per se, >>> other than the filter sets that may be recommended for use with them. >>> >>> Jeff >>> >>> *Jeff Carmichael* >>> *Product & Technical Marketing Manager* >>> >>> *[hidden email] <[hidden email]> | 802-428-2528* >>> >>> * <[hidden email]>* >>> >> -- George McNamara, PhD Baltimore, MD 21231 [hidden email] https://www.linkedin.com/in/georgemcnamara https://works.bepress.com/gmcnamara/75 (may need to use Microsoft Edge or Firefox, rather than Google Chrome) http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650 http://confocal.jhu.edu July 2017 Current Protocols article, open access: UNIT 4.4 Microscopy and Image Analysis http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/abstract supporting materials direct link is http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/full#hg0404-sec-0023 figures at http://onlinelibrary.wiley.com/doi/10.1002/cphg.42/figures |
|
Jeffrey Carmichael |
|
|
In reply to this post by Sripad Ram-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Sripad, I don't recall seeing a direct comparison of image data with TSA-prepared samples either, but I'd hesitate to comment on that given that I haven't performed this processing myself. The TSA protocols I'm familiar with involve sequential labeling steps involving stripping by microwaving tissue samples between each successive round, and overall involve a very large number of total processing steps. I can't help but be skeptical of these protocols. I know there are multiple TSA protocols, but I'm not very familiar with them. Part of what intuitively appeals to me about this peptide-hapten approach is it's simplicity and relative ease, which I would expect to aid reproducibility, especially when also considering the high specificity of the secondary reagents. I have seen data that I'm not at liberty to disclose, and I also lack very detailed knowledge about the technology to discuss intelligently, so I'd defer to David Schwartz at CellIDx to respond. Jeff On Thu, Aug 23, 2018 at 9:44 PM, S Ram <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Jeff, > My understanding is that cellIdx technology can support upto a 4plex (i.e. > 4 abs). > > I dont recall seeing side by side comparison of their technology with TSA, > but what i hear is that the signal intensity is good enough to detect Ab > labeling even in notoriously autofluorescent samples (lung cancer tissue). > > Sripad > > > On Thu, Aug 23, 2018, 2:02 PM Jeff Carmichael <[hidden email]> > wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > COMMERCIAL POST > > > > In a recent conversation thread about blue fluorochromes, George McNamara > > mentioned alternative labeling methods such as biotin/streptavidin and > > TSA-hapten. > > > > This reminded me of a customer we've been working with who has developed > > what they're calling a "next generation" hapten (not TSA) labeling > > technology with very high specificity, affinity and amplification. This > in > > turn means that the primary antibody species origin doesn't matter, and 4 > > mouse monoclonal antibodies may be used together on the same tissue > sample, > > including without multiple labeling steps. The "modified" haptens are > > referred to as peptide bar-codes which apparently allow for significant > > multiplexing: > > > > https://cellidx.com/technology > > https://event.webcasts.com/viewer/event.jsp?ei=1187871&tp_key=9db65f5892 > > > > Chroma Technology has no commercial interest in these products per se, > > other than the filter sets that may be recommended for use with them. > > > > Jeff > > > > *Jeff Carmichael* > > *Product & Technical Marketing Manager* > > > > *[hidden email] <[hidden email]> | 802-428-2528* > > > > * <[hidden email]>* > > > > -- > > <https://www.chroma.com/>CHROMA TECHNOLOGY CORP® > > *an employee owned > > company* > > 10 Imtec Lane, Bellows Falls, Vermont 05101 USA > > 800-824-7662 | > > FAX: 802-428-2525 > > www.chroma.com <https://www.chroma.com/> | > > [hidden email] <mailto:[hidden email]> > > > -- <https://www.chroma.com/>CHROMA TECHNOLOGY CORP® *an employee owned company* 10 Imtec Lane, Bellows Falls, Vermont 05101 USA 800-824-7662 | FAX: 802-428-2525 www.chroma.com <https://www.chroma.com/> | [hidden email] <mailto:[hidden email]> |
|
Jim Mansfield |
|
|
In reply to this post by Jeffrey Carmichael
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi, I have heard very good things about Cell IDx's multiplexing methodology, so I am glad to hear that word of it is spreading. I'd love to hear if anyone has any hands-on experience comparing it to TSA (either PerkinElmer's Opal or Roche's TSA variant). But since this thread is partly about multiplex IHC methods, I thought I'd mention Ultivue, who has developed their own methodology for doing multiplex IHC staining: https://www.ultivue.com/ I have not tried either Cell IDx or Ultivue, so I can't give any comments on how they compare - but I'd love to hear from someone who has! Best, Jim Andor Technology Commercial interest: I have no commercial interest in any of these methods ... although in the spirit of full disclosure, I used to work at PerkinElmer in their Quantitative Pathology group and am very familiar with the TSA methods. Jim Mansfield [hidden email] -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Jeff Carmichael Sent: Thursday, August 23, 2018 3:57 PM To: [hidden email] Subject: Immunofluorescence multiplexing with an interesting "modified hapten" labeling technology ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** COMMERCIAL POST In a recent conversation thread about blue fluorochromes, George McNamara mentioned alternative labeling methods such as biotin/streptavidin and TSA-hapten. This reminded me of a customer we've been working with who has developed what they're calling a "next generation" hapten (not TSA) labeling technology with very high specificity, affinity and amplification. This in turn means that the primary antibody species origin doesn't matter, and 4 mouse monoclonal antibodies may be used together on the same tissue sample, including without multiple labeling steps. The "modified" haptens are referred to as peptide bar-codes which apparently allow for significant multiplexing: https://cellidx.com/technology https://event.webcasts.com/viewer/event.jsp?ei=1187871&tp_key=9db65f5892 Chroma Technology has no commercial interest in these products per se, other than the filter sets that may be recommended for use with them. Jeff *Jeff Carmichael* *Product & Technical Marketing Manager* *[hidden email] <[hidden email]> | 802-428-2528* * <[hidden email]>* -- <https://www.chroma.com/>CHROMA TECHNOLOGY CORP® *an employee owned company* 10 Imtec Lane, Bellows Falls, Vermont 05101 USA 800-824-7662 | FAX: 802-428-2525 www.chroma.com <https://www.chroma.com/> | [hidden email] <mailto:[hidden email]> |
|
Sripad Ram-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Jim, Glad to note that you are mentioning Ultivue. On the same spirit, we should also mention Akoya biosciences which is a spinoff of Gary Nolan's CODEX technology. Both use dna labeled primary Abs with slight variation in washing strategies. As for a direct comparison between Ultivue and Perkin Elmer, that is something that we are potentially interested and may pursue at some point in time. Jeff, I am very familiar with perkin elmer Opal dyes and have been using them for well over 2 years. There are ways you can do Ab removal without doing microwave treatment. However you'd need an auto stainer for this (such as a leica bond rx or similar device). Opal dyes have their challenges but its not difficult to have the right controls to check for tissue loss/damage. The bigger challenges are with spectral unmixing and the downstream image analysis pipelines. Sripad |
|
Jeffrey Carmichael |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Sripad and Jim, Yes, the DNA barcoding approach is potentially very powerful. I'm not very familiar with it but I see this paper, published this month in Cell: https://www.cell.com/cell/pdf/S0092-8674(18)30904-8.pdf Otherwise, my experience is that simplicity will aid reproducibility. That's partly what is attractive to me about the modified-hapten approach. It sounds like you have more patience than I do regarding complex sample processing! I'm curious what the utility of these various techniques will be in applications ranging from basic research to preclinical to highly multiplexed clinical data acquisition. All have somewhat different aims & requirements. Jeff On Sat, Aug 25, 2018 at 1:57 AM, S Ram <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Jim, > Glad to note that you are mentioning Ultivue. On the same spirit, we should > also mention Akoya biosciences which is a spinoff of Gary Nolan's CODEX > technology. Both use dna labeled primary Abs with slight variation in > washing strategies. As for a direct comparison between Ultivue and Perkin > Elmer, that is something that we are potentially interested and may pursue > at some point in time. > > Jeff, > I am very familiar with perkin elmer Opal dyes and have been using them for > well over 2 years. There are ways you can do Ab removal without doing > microwave treatment. However you'd need an auto stainer for this (such as a > leica bond rx or similar device). > Opal dyes have their challenges but its not difficult to have the right > controls to check for tissue loss/damage. The bigger challenges are with > spectral unmixing and the downstream image analysis pipelines. > > Sripad > -- <https://www.chroma.com/>CHROMA TECHNOLOGY CORP® *an employee owned company* 10 Imtec Lane, Bellows Falls, Vermont 05101 USA 800-824-7662 | FAX: 802-428-2525 www.chroma.com <https://www.chroma.com/> | [hidden email] <mailto:[hidden email]> |