|
Ekaterina PAPUSHEVA-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, What is your experience with Jobs software module that Nikon offers? we are thinking of using it with a widefield microscope. On a paper it looks really promising, so I wonder what are the experiences of real users? All best, Katja ___ Ekaterina Papusheva Manager of Bioimaging Facility Institute of Science and Technology Austria Phone +43-(0)2243 9000-1043; Mobile +43-(0)664 88509130 Fax +43-(0)2243 9000-2000 Am Campus 1 A-3400 Klosterneuburg Visit our website: www.ist.ac.at<http://www.ist.ac.at> This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. |
|
simon walker (BI)-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Katja, We've been using JOBS on a Nikon wide-field system for a year or so now. The more I use it, the more impressed I am. It has a very user-friendly interface, a huge amount of functionality and is generally stable. It allows you to convert your Nikon microscope into what is essentially a high content imaging device, integrating complex image capture routines with analysis which can then be used to direct further image acquisition if required. It wouldn't necessarily replace an HCS system, but it has more functionality than many bespoke HC imagers. I recently added JOBS to our A1 confocal to enable some of the functionality that we were using on the wide-field system on that platform. Simon (no commercial interest) -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ekaterina PAPUSHEVA Sent: 19 October 2017 12:40 To: [hidden email] Subject: JOBS software module from Nikon ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, What is your experience with Jobs software module that Nikon offers? we are thinking of using it with a widefield microscope. On a paper it looks really promising, so I wonder what are the experiences of real users? All best, Katja ___ Ekaterina Papusheva Manager of Bioimaging Facility Institute of Science and Technology Austria Phone +43-(0)2243 9000-1043; Mobile +43-(0)664 88509130 Fax +43-(0)2243 9000-2000 Am Campus 1 A-3400 Klosterneuburg Visit our website: www.ist.ac.at<http://www.ist.ac.at> This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. The Babraham Institute, Babraham Research Campus, Cambridge CB22 3AT Registered Charity No. 1053902. The information transmitted in this email is directed only to the addressee. If you received this in error, please contact the sender and delete this email from your system. The contents of this e-mail are the views of the sender and do not necessarily represent the views of the Babraham Institute. Full conditions at: www.babraham.ac.uk<http://www.babraham.ac.uk/terms> |
|
Gary Laevsky |
|
|
In reply to this post by Ekaterina PAPUSHEVA-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Katja, Works really well for us, particularly when coupled with General Analysis (GA). Almost like, or is, machine learning. Best, Gary > On Oct 19, 2017, at 7:39 AM, Ekaterina PAPUSHEVA <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi all, > > What is your experience with Jobs software module that Nikon offers? we are thinking of using it with a widefield microscope. > On a paper it looks really promising, so I wonder what are the experiences of real users? > > All best, > Katja > > ___ > > Ekaterina Papusheva > > Manager of Bioimaging Facility > > Institute of Science and Technology Austria > > Phone +43-(0)2243 9000-1043; > Mobile +43-(0)664 88509130 > Fax +43-(0)2243 9000-2000 > > Am Campus 1 > > A-3400 Klosterneuburg > > Visit our website: www.ist.ac.at<http://www.ist.ac.at> > > > This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. |
|
Csúcs Gábor-3 |
|
|
In reply to this post by Ekaterina PAPUSHEVA-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Katja, We also had very good experience with it - it offers a very flexible platform. Some users may be a little bit overwhelmed by the possibilities but for experts it is a powerful option. Greetings Gabor -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ekaterina PAPUSHEVA Sent: Thursday, October 19, 2017 1:40 PM To: [hidden email] Subject: JOBS software module from Nikon ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, What is your experience with Jobs software module that Nikon offers? we are thinking of using it with a widefield microscope. On a paper it looks really promising, so I wonder what are the experiences of real users? All best, Katja ___ Ekaterina Papusheva Manager of Bioimaging Facility Institute of Science and Technology Austria Phone +43-(0)2243 9000-1043; Mobile +43-(0)664 88509130 Fax +43-(0)2243 9000-2000 Am Campus 1 A-3400 Klosterneuburg Visit our website: www.ist.ac.at<http://www.ist.ac.at> This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. |
|
giuseppe sancataldo |
|
|
In reply to this post by Gary Laevsky
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Katja I had the chance to use JOBS during my PhD. It is possible to create arbitrary procedures in a really easy way (users only need to be familiar with simple programming concepts). My opinion about JOBS is positive. Best, GS -- Giuseppe Sancataldo, PhD LENS - European Laboratory for Non-Linear Spectroscopy University of Florence Via Nello Carrara 1 - 50019 Sesto Fiorentino (Firenze) - Italy Mobile: +39 333 9477002 email: [hidden email] Il 19/10/2017 14:14, Gary Laevsky ha scritto: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Katja, > > Works really well for us, particularly when coupled with General Analysis (GA). Almost like, or is, machine learning. > > Best, > > Gary > > > >> On Oct 19, 2017, at 7:39 AM, Ekaterina PAPUSHEVA <[hidden email]> wrote: >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Hi all, >> >> What is your experience with Jobs software module that Nikon offers? we are thinking of using it with a widefield microscope. >> On a paper it looks really promising, so I wonder what are the experiences of real users? >> >> All best, >> Katja >> >> ___ >> >> Ekaterina Papusheva >> >> Manager of Bioimaging Facility >> >> Institute of Science and Technology Austria >> >> Phone +43-(0)2243 9000-1043; >> Mobile +43-(0)664 88509130 >> Fax +43-(0)2243 9000-2000 >> >> Am Campus 1 >> >> A-3400 Klosterneuburg >> >> Visit our website: www.ist.ac.at<http://www.ist.ac.at> >> >> >> This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. > --- Questa e-mail è stata controllata per individuare virus con Avast antivirus. https://www.avast.com/antivirus |
|
Sylvie Le Guyader |
|
|
In reply to this post by Ekaterina PAPUSHEVA-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Work a charm for us. We love imaging fast a large area in multiwell plates, analysing to find cells or regions of interest and revisiting these points to re image with a z stack, at higher res and more channels for example. Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader, PhD Live Cell Imaging Facility Manager Karolinska Institutet- Bionut Dpt Hälsovägen 7, Novum, G lift, floor 6 14157 Huddinge Sweden mobile: +46 (0) 73 733 5008 office: +46 (0) 8 524 811 72 LCI website -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ekaterina PAPUSHEVA Sent: den 19 oktober 2017 13:40 To: [hidden email] Subject: JOBS software module from Nikon ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, What is your experience with Jobs software module that Nikon offers? we are thinking of using it with a widefield microscope. On a paper it looks really promising, so I wonder what are the experiences of real users? All best, Katja ___ Ekaterina Papusheva Manager of Bioimaging Facility Institute of Science and Technology Austria Phone +43-(0)2243 9000-1043; Mobile +43-(0)664 88509130 Fax +43-(0)2243 9000-2000 Am Campus 1 A-3400 Klosterneuburg Visit our website: www.ist.ac.at<http://www.ist.ac.at> This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. |
|
Mark A. Sanders |
|
|
In reply to this post by Ekaterina PAPUSHEVA-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Greetings, As a Nikon Center of Excellence we have a growing fan base of JOBS from basic automation to fairly sophisticated multi-pass imaging based on analysis criteria especially on our A1R confocal and Ti widefield scopes. We have numerous satisfied users. There is also a growing interest in creating shareable libraries of experiments to help streamline the adoption of the JOBS workflow for new users. Hopefully that will help novice users learn the workflow faster. Cheers, Mark -- Mark Sanders University of Minnesota Program Director Twin Cities Campus University Imaging Centers Nikon Center of Excellence uic.umn.edu uic.umn.edu/content/locations Any work utilizing UIC equipment or staff support must acknowledge the University Imaging Centers in publications and presentations. > On Oct 19, 2017, at 6:39 AM, Ekaterina PAPUSHEVA <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi all, > > What is your experience with Jobs software module that Nikon offers? we are thinking of using it with a widefield microscope. > On a paper it looks really promising, so I wonder what are the experiences of real users? > > All best, > Katja > > ___ > > Ekaterina Papusheva > > Manager of Bioimaging Facility > > Institute of Science and Technology Austria > > Phone +43-(0)2243 9000-1043; > Mobile +43-(0)664 88509130 > Fax +43-(0)2243 9000-2000 > > Am Campus 1 > > A-3400 Klosterneuburg > > Visit our website: www.ist.ac.at<http://www.ist.ac.at> > > > This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. |
|
Michael Abanto |
|
|
In reply to this post by Csúcs Gábor-3
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I use it almost daily to automate slide and multi-well scanning in our facility. It's great but requires some time to learn (faster to learn and use than e.g. python but you can hit the ceiling a bit earlier (though you can combine with macros)). There are a few limitations (e.g. size and saving of initial scan), but it is quite powerful and flexible. Most of our users (mainly bio or medical) dont learn it, but you can make e.g. wizards for their JOB. I'm a fan and would recommend it. You can contact me off list if needed, Mike On Thu, Oct 19, 2017 at 2:20 PM Csúcs Gábor <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear Katja, > > We also had very good experience with it - it offers a very flexible > platform. Some users may be a little bit overwhelmed by the possibilities > but for experts it is a powerful option. > > Greetings Gabor > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Ekaterina PAPUSHEVA > Sent: Thursday, October 19, 2017 1:40 PM > To: [hidden email] > Subject: JOBS software module from Nikon > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi all, > > What is your experience with Jobs software module that Nikon offers? we > are thinking of using it with a widefield microscope. > On a paper it looks really promising, so I wonder what are the experiences > of real users? > > All best, > Katja > > ___ > > Ekaterina Papusheva > > Manager of Bioimaging Facility > > Institute of Science and Technology Austria > > Phone +43-(0)2243 9000-1043; > Mobile +43-(0)664 88509130 <+43%20664%2088509130> > Fax +43-(0)2243 9000-2000 > > Am Campus 1 <https://maps.google.com/?q=Am+Campus+1&entry=gmail&source=g> > > A-3400 Klosterneuburg > > Visit our website: www.ist.ac.at<http://www.ist.ac.at> > > > This message and any attachment (together "the Message") are confidential > and intended only for the individual named. If you have received the > Message in error you should not distribute or copy this Message. Please > notify the sender immediately by e-mail if you have received this Message > in error and delete the Message from your system. Correspondence via e-mail > is for information purposes. IST Austria neither makes nor accepts legally > binding statements unless explicitly otherwise agreed to the contrary. > |
|
Ekaterina PAPUSHEVA-2 |
|
|
In reply to this post by Sylvie Le Guyader
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thanks everyone, I think we go for it, as we would really profit from advanced jobs, like object tracking or decision-making on the basis of intensity. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Sylvie Le Guyader Sent: Thursday, October 19, 2017 2:29 PM To: [hidden email] Subject: Re: JOBS software module from Nikon ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Work a charm for us. We love imaging fast a large area in multiwell plates, analysing to find cells or regions of interest and revisiting these points to re image with a z stack, at higher res and more channels for example. Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader, PhD Live Cell Imaging Facility Manager Karolinska Institutet- Bionut Dpt Hälsovägen 7, Novum, G lift, floor 6 14157 Huddinge Sweden mobile: +46 (0) 73 733 5008 office: +46 (0) 8 524 811 72 LCI website -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ekaterina PAPUSHEVA Sent: den 19 oktober 2017 13:40 To: [hidden email] Subject: JOBS software module from Nikon ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, What is your experience with Jobs software module that Nikon offers? we are thinking of using it with a widefield microscope. On a paper it looks really promising, so I wonder what are the experiences of real users? All best, Katja ___ Ekaterina Papusheva Manager of Bioimaging Facility Institute of Science and Technology Austria Phone +43-(0)2243 9000-1043; Mobile +43-(0)664 88509130 Fax +43-(0)2243 9000-2000 Am Campus 1 A-3400 Klosterneuburg Visit our website: www.ist.ac.at<http://www.ist.ac.at> This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. |
|
PEARSON Matthew |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, We have just purchased JOBS and so far pretty impressed. I wonder if a JOBS user forum or email list exists where you could ask advice from the JOBS user community? We have had good support from Nikon so far but the people learning the most about what it can do are probably the facility staff/users who have time to sit on the microscopes. Might be nice if this was official and could be created and administered by Nikon, I may ask our local reps about this. Thanks, Matt -- Matt Pearson Microscopy Facility MRC Human Genetics Unit Institute of Genetics and Molecular Medicine (IGMM) University of Edinburgh Crewe Road EH4 2XU On 19 Oct 2017, at 15:49, Ekaterina PAPUSHEVA <[hidden email]<mailto:[hidden email]>> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thanks everyone, I think we go for it, as we would really profit from advanced jobs, like object tracking or decision-making on the basis of intensity. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Sylvie Le Guyader Sent: Thursday, October 19, 2017 2:29 PM To: [hidden email] Subject: Re: JOBS software module from Nikon ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Work a charm for us. We love imaging fast a large area in multiwell plates, analysing to find cells or regions of interest and revisiting these points to re image with a z stack, at higher res and more channels for example. Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader, PhD Live Cell Imaging Facility Manager Karolinska Institutet- Bionut Dpt Hälsovägen 7, Novum, G lift, floor 6 14157 Huddinge Sweden mobile: +46 (0) 73 733 5008 office: +46 (0) 8 524 811 72 LCI website -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ekaterina PAPUSHEVA Sent: den 19 oktober 2017 13:40 To: [hidden email] Subject: JOBS software module from Nikon ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, What is your experience with Jobs software module that Nikon offers? we are thinking of using it with a widefield microscope. On a paper it looks really promising, so I wonder what are the experiences of real users? All best, Katja ___ Ekaterina Papusheva Manager of Bioimaging Facility Institute of Science and Technology Austria Phone +43-(0)2243 9000-1043; Mobile +43-(0)664 88509130 Fax +43-(0)2243 9000-2000 Am Campus 1 A-3400 Klosterneuburg Visit our website: www.ist.ac.at<http://www.ist.ac.at> This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** The University of Edinburgh is a charitable body, registered in Scotland, with registration number SC005336. |
|
Christian Elowsky-2 |
|
|
In reply to this post by Ekaterina PAPUSHEVA-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Apologies for the late response to this thread. I was given a walkthrough this week for JOBS, and I'm certainly impressed with the flexibility and potential. While I understood the basics of building recipes (as they call them), I am looking for more instruction. Do any of the labs using JOBS have anything to offer along these lines? Our core images everything that walks through the door, so simple set-ups will not quite fit the bill. Anything will be helpful as we begin writing an in-house tutorial. Thanks. Christian ________________________________ From: Confocal Microscopy List <[hidden email]> on behalf of Ekaterina PAPUSHEVA <[hidden email]> Sent: Thursday, October 19, 2017 9:49:55 AM To: [hidden email] Subject: Re: JOBS software module from Nikon ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thanks everyone, I think we go for it, as we would really profit from advanced jobs, like object tracking or decision-making on the basis of intensity. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Sylvie Le Guyader Sent: Thursday, October 19, 2017 2:29 PM To: [hidden email] Subject: Re: JOBS software module from Nikon ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Work a charm for us. We love imaging fast a large area in multiwell plates, analysing to find cells or regions of interest and revisiting these points to re image with a z stack, at higher res and more channels for example. Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader, PhD Live Cell Imaging Facility Manager Karolinska Institutet- Bionut Dpt Hälsovägen 7, Novum, G lift, floor 6 14157 Huddinge Sweden mobile: +46 (0) 73 733 5008 office: +46 (0) 8 524 811 72 LCI website -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ekaterina PAPUSHEVA Sent: den 19 oktober 2017 13:40 To: [hidden email] Subject: JOBS software module from Nikon ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, What is your experience with Jobs software module that Nikon offers? we are thinking of using it with a widefield microscope. On a paper it looks really promising, so I wonder what are the experiences of real users? All best, Katja ___ Ekaterina Papusheva Manager of Bioimaging Facility Institute of Science and Technology Austria Phone +43-(0)2243 9000-1043; Mobile +43-(0)664 88509130 Fax +43-(0)2243 9000-2000 Am Campus 1 A-3400 Klosterneuburg Visit our website: www.ist.ac.at<http://www.ist.ac.at> This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. |
|
Sylvie Le Guyader |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Christian Nothing about Jobs itself but a little video to get people started with General Analysis which is part of jobs. Jobs realizes its full potential when several layers of acquisition /analysis /re acquisition are embedded in the pipeline. The design of GA has changed in NIS 5 but the principles remain. There are a few more videos, some about microscopy in general and some about NIS.boîte you will find them useful. Look under Learning material http://ki.se/en/bionut/learning-microscopy Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader, PhD Live Cell Imaging Facility Manager Karolinska Institutet- Bionut Dpt Hälsovägen 7, Novum, G lift, floor 6 14157 Huddinge Sweden mobile: +46 (0) 73 733 5008 office: +46 (0) 8 524 811 72 LCI website ---- Christian Elowsky wrote ---- ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Apologies for the late response to this thread. I was given a walkthrough this week for JOBS, and I'm certainly impressed with the flexibility and potential. While I understood the basics of building recipes (as they call them), I am looking for more instruction. Do any of the labs using JOBS have anything to offer along these lines? Our core images everything that walks through the door, so simple set-ups will not quite fit the bill. Anything will be helpful as we begin writing an in-house tutorial. Thanks. Christian ________________________________ From: Confocal Microscopy List <[hidden email]> on behalf of Ekaterina PAPUSHEVA <[hidden email]> Sent: Thursday, October 19, 2017 9:49:55 AM To: [hidden email] Subject: Re: JOBS software module from Nikon ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thanks everyone, I think we go for it, as we would really profit from advanced jobs, like object tracking or decision-making on the basis of intensity. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Sylvie Le Guyader Sent: Thursday, October 19, 2017 2:29 PM To: [hidden email] Subject: Re: JOBS software module from Nikon ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Work a charm for us. We love imaging fast a large area in multiwell plates, analysing to find cells or regions of interest and revisiting these points to re image with a z stack, at higher res and more channels for example. Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader, PhD Live Cell Imaging Facility Manager Karolinska Institutet- Bionut Dpt Hälsovägen 7, Novum, G lift, floor 6 14157 Huddinge Sweden mobile: +46 (0) 73 733 5008 office: +46 (0) 8 524 811 72 LCI website -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ekaterina PAPUSHEVA Sent: den 19 oktober 2017 13:40 To: [hidden email] Subject: JOBS software module from Nikon ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, What is your experience with Jobs software module that Nikon offers? we are thinking of using it with a widefield microscope. On a paper it looks really promising, so I wonder what are the experiences of real users? All best, Katja ___ Ekaterina Papusheva Manager of Bioimaging Facility Institute of Science and Technology Austria Phone +43-(0)2243 9000-1043; Mobile +43-(0)664 88509130 Fax +43-(0)2243 9000-2000 Am Campus 1 A-3400 Klosterneuburg Visit our website: www.ist.ac.at<http://www.ist.ac.at> This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. |
|
Christian Elowsky-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Sylvie, Thank you for the information. I'm looking forward to understanding the full potential. It appears that it may be a lengthy process. Christian From: Sylvie Le Guyader Sent: Friday, October 27, 1:43 AM Subject: Re: JOBS software module from Nikon To: [hidden email] ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Christian Nothing about Jobs itself but a little video to get people started with General Analysis which is part of jobs. Jobs realizes its full potential when several layers of acquisition /analysis /re acquisition are embedded in the pipeline. The design of GA has changed in NIS 5 but the principles remain. There are a few more videos, some about microscopy in general and some about NIS.boîte you will find them useful. Look under Learning material http://ki.se/en/bionut/learning-microscopy Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader, PhD Live Cell Imaging Facility Manager Karolinska Institutet- Bionut Dpt Hälsovägen 7, Novum, G lift, floor 6 14157 Huddinge Sweden mobile: +46 (0) 73 733 5008 office: +46 (0) 8 524 811 72 LCI website ---- Christian Elowsky wrote ---- ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Apologies for the late response to this thread. I was given a walkthrough this week for JOBS, and I'm certainly impressed with the flexibility and potential. While I understood the basics of building recipes (as they call them), I am looking for more instruction. Do any of the labs using JOBS have anything to offer along these lines? Our core images everything that walks through the door, so simple set-ups will not quite fit the bill. Anything will be helpful as we begin writing an in-house tutorial. Thanks. Christian ________________________________ From: Confocal Microscopy List on behalf of Ekaterina PAPUSHEVA Sent: Thursday, October 19, 2017 9:49:55 AM To: [hidden email] Subject: Re: JOBS software module from Nikon ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thanks everyone, I think we go for it, as we would really profit from advanced jobs, like object tracking or decision-making on the basis of intensity. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Sylvie Le Guyader Sent: Thursday, October 19, 2017 2:29 PM To: [hidden email] Subject: Re: JOBS software module from Nikon ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Work a charm for us. We love imaging fast a large area in multiwell plates, analysing to find cells or regions of interest and revisiting these points to re image with a z stack, at higher res and more channels for example. Med vänlig hälsning / Best regards Sylvie @@@@@@@@@@@@@@@@@@@@@@@@ Sylvie Le Guyader, PhD Live Cell Imaging Facility Manager Karolinska Institutet- Bionut Dpt Hälsovägen 7, Novum, G lift, floor 6 14157 Huddinge Sweden mobile: +46 (0) 73 733 5008 office: +46 (0) 8 524 811 72 LCI website -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ekaterina PAPUSHEVA Sent: den 19 oktober 2017 13:40 To: [hidden email] Subject: JOBS software module from Nikon ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, What is your experience with Jobs software module that Nikon offers? we are thinking of using it with a widefield microscope. On a paper it looks really promising, so I wonder what are the experiences of real users? All best, Katja ___ Ekaterina Papusheva Manager of Bioimaging Facility Institute of Science and Technology Austria Phone +43-(0)2243 9000-1043; Mobile +43-(0)664 88509130 Fax +43-(0)2243 9000-2000 Am Campus 1 A-3400 Klosterneuburg Visit our website: www.ist.ac.at This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. |
| Free forum by Nabble | Edit this page |