To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** We have a user who wants to kill specific GFP labelled single cells in a zebra fish and follow the developmental consequences. Multiphoton or confocal used to zap a selected area was the suggestion but Is bleaching the GFP sufficient to damage the cell. How can we determine if the cell is dead/damaged - trypan blue in fish. The extent to which surrounding cells are also damaged. Are there any photoactivatable toxins ? Ideas, references, suggestions would be greatly appreciated. Jeremy Adler =============================================== B i o V i s P l a t f o r m of Uppsala University Light & EM microscopy / FlowCytometry & Cell Sorting / Image Analysis =============================================== Jeremy Adler PhD - Senior research engineer Light, Confocal Microscopy, Image Analysis E-mail: [hidden email] 070-1679349 Dag Hammarskjölds v 20 751 85 UPPSALA, SWEDEN http://biovis.uu.se/ =============================================== När du har kontakt med oss på Uppsala universitet med e-post så innebär det att vi behandlar dina personuppgifter. För att läsa mer om hur vi gör det kan du läsa här: http://www.uu.se/om-uu/dataskydd-personuppgifter/ E-mailing Uppsala University means that we will process your personal data. For more information on how this is performed, please read here: http://www.uu.se/en/about-uu/data-protection-policy |
![]() ![]() |
Rowlands, Christopher J |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Photodynamic therapy does exactly this - I've used verteporfin in the past (http://dx.doi.org/10.1088/1612-2011/11/11/115605), but there's plenty of alternative PDT agents. Regarding the assessment of the cell, sacrifice the fish and LIVE/DEAD stain? Or if you need it still alive, bleach the GFP and see whether the GFP intensity recovers over time (suggesting that protein synthesis is still occurring). Chris -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of Jeremy Adler Sent: 15 August 2019 12:41 To: [hidden email] Subject: Killing cells in a zebra fish ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** We have a user who wants to kill specific GFP labelled single cells in a zebra fish and follow the developmental consequences. Multiphoton or confocal used to zap a selected area was the suggestion but Is bleaching the GFP sufficient to damage the cell. How can we determine if the cell is dead/damaged - trypan blue in fish. The extent to which surrounding cells are also damaged. Are there any photoactivatable toxins ? Ideas, references, suggestions would be greatly appreciated. Jeremy Adler =============================================== B i o V i s P l a t f o r m of Uppsala University Light & EM microscopy / FlowCytometry & Cell Sorting / Image Analysis =============================================== Jeremy Adler PhD - Senior research engineer Light, Confocal Microscopy, Image Analysis E-mail: [hidden email] 070-1679349 Dag Hammarskjölds v 20 751 85 UPPSALA, SWEDEN http://biovis.uu.se/ =============================================== När du har kontakt med oss på Uppsala universitet med e-post så innebär det att vi behandlar dina personuppgifter. För att läsa mer om hur vi gör det kan du läsa här: http://www.uu.se/om-uu/dataskydd-personuppgifter/ E-mailing Uppsala University means that we will process your personal data. For more information on how this is performed, please read here: http://www.uu.se/en/about-uu/data-protection-policy |
![]() ![]() |
bioimaging@uni-konstanz.de |
![]() |
In reply to this post by Jeremy Adler-4
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Jeremy, you might consider using "Killer-GFPs" like Killer-Red or Supernova. They are used for CALI, chromophore assisted laser inactivation. If the cytoplasm is full of it, then the iiradiated cell should die. You might be able to use really low laser powers. I don´t know about fish, but in culture cells Calcein is a good life/dead stain. Best Elisa On Thu, Aug 15, 2019 at 1:42 PM Jeremy Adler <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > > We have a user who wants to kill specific GFP labelled single cells in a > zebra fish and follow the developmental consequences. > Multiphoton or confocal used to zap a selected area was the suggestion but > > Is bleaching the GFP sufficient to damage the cell. > How can we determine if the cell is dead/damaged - trypan blue in fish. > The extent to which surrounding cells are also damaged. > Are there any photoactivatable toxins ? > > Ideas, references, suggestions would be greatly appreciated. > > > Jeremy Adler > =============================================== > B i o V i s P l a t f o r m of Uppsala University > Light & EM microscopy / FlowCytometry & Cell Sorting / > Image Analysis =============================================== > Jeremy Adler PhD - Senior research engineer Light, Confocal Microscopy, > Image Analysis > E-mail: [hidden email] > 070-1679349 > > Dag Hammarskjölds v 20 > 751 85 UPPSALA, SWEDEN > http://biovis.uu.se/ > =============================================== > > > > > > > > > > > > > > > > > När du har kontakt med oss på Uppsala universitet med e-post så innebär > det att vi behandlar dina personuppgifter. För att läsa mer om hur vi gör > det kan du läsa här: http://www.uu.se/om-uu/dataskydd-personuppgifter/ > > E-mailing Uppsala University means that we will process your personal > data. For more information on how this is performed, please read here: > http://www.uu.se/en/about-uu/data-protection-policy > -- Bioimaging Center University of Konstanz P.O. Box 604 D-78457 Konstanz Germany Tel. +49-7531-884054 [hidden email] http.//www.uni-konstanz.de/bioimaging |
![]() ![]() |
Tim Feinstein |
![]() |
In reply to this post by Jeremy Adler-4
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Jeremy, I'd recommend using a mitochondrial or nuclear-targeted KillerRed for that purpose. You'd need a pretty modest 555nm light flux to zap targeted cells into apoptosis. There are probably newer and better genetically targeted photosensitizing agents, but that's the one that I've used. Best, T Timothy Feinstein, Ph.D. Research Scientist Department of Developmental Biology University of Pittsburgh On 8/15/19, 7:42 AM, "Confocal Microscopy List on behalf of Jeremy Adler" <[hidden email] on behalf of [hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: https://nam05.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7Ctnf8%40PITT.EDU%7Cf35f8011aca548894d7108d721759a93%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637014661334143515&sdata=pX1oV6q0ojLWNiot%2B6k3cF%2BSa%2B1CIgp6kBFFqD3vDhY%3D&reserved=0 Post images on https://nam05.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=02%7C01%7Ctnf8%40PITT.EDU%7Cf35f8011aca548894d7108d721759a93%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637014661334143515&sdata=2oGrBtdIHgZ7WBT04bL9GfVCr9Ce9rQhkNIVgzUPdp4%3D&reserved=0 and include the link in your posting. ***** We have a user who wants to kill specific GFP labelled single cells in a zebra fish and follow the developmental consequences. Multiphoton or confocal used to zap a selected area was the suggestion but Is bleaching the GFP sufficient to damage the cell. How can we determine if the cell is dead/damaged - trypan blue in fish. The extent to which surrounding cells are also damaged. Are there any photoactivatable toxins ? Ideas, references, suggestions would be greatly appreciated. Jeremy Adler =============================================== B i o V i s P l a t f o r m of Uppsala University Light & EM microscopy / FlowCytometry & Cell Sorting / Image Analysis =============================================== Jeremy Adler PhD - Senior research engineer Light, Confocal Microscopy, Image Analysis E-mail: [hidden email] 070-1679349 Dag Hammarskjölds v 20 751 85 UPPSALA, SWEDEN https://nam05.safelinks.protection.outlook.com/?url=http%3A%2F%2Fbiovis.uu.se%2F&data=02%7C01%7Ctnf8%40PITT.EDU%7Cf35f8011aca548894d7108d721759a93%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637014661334143515&sdata=n0bjom4QhdtLqoa0wFb3aaaAfUsgq0t%2FNQa%2BdWZbGA4%3D&reserved=0 =============================================== När du har kontakt med oss på Uppsala universitet med e-post så innebär det att vi behandlar dina personuppgifter. För att läsa mer om hur vi gör det kan du läsa här: https://nam05.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.uu.se%2Fom-uu%2Fdataskydd-personuppgifter%2F&data=02%7C01%7Ctnf8%40PITT.EDU%7Cf35f8011aca548894d7108d721759a93%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637014661334143515&sdata=YPoOYRsJhBDx8rp%2BEYeet3KQjNcU32wsE%2BBF0jOqCXk%3D&reserved=0 E-mailing Uppsala University means that we will process your personal data. For more information on how this is performed, please read here: https://nam05.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.uu.se%2Fen%2Fabout-uu%2Fdata-protection-policy&data=02%7C01%7Ctnf8%40PITT.EDU%7Cf35f8011aca548894d7108d721759a93%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637014661334143515&sdata=nKa%2FEM0NxEJqxUJNiadUbCSxJGHAq3dDh2djGjIKkAE%3D&reserved=0 |
![]() ![]() |
Drummond, Iain A. |
![]() |
In reply to this post by Jeremy Adler-4
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Yes we have photosensitized cells with GFP. We used a 405 laser line to ablate pronephric kidney tubules. You can use propidium iodide to track the cell death. Here's one of our papers on this: https://www.ncbi.nlm.nih.gov/pmc/articles/PMC4092191/ -- Iain Drummond, PhD Associate Professor, Nephrology Division Massachusetts General Hospital 149 13th Street Charlestown MA 02129 Department of Genetics, Program in Developmental and Regenerative Biology, Harvard Medical School Director, Harvard Stem Cell Institute Kidney Program 617 726 5647 617 726 5669 (fax) [hidden email] [hidden email] http://drummondlab.mgh.harvard.edu On 8/15/19, 7:42 AM, "Confocal Microscopy List on behalf of Jeremy Adler" <[hidden email] on behalf of [hidden email]> wrote: External Email - Use Caution ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** We have a user who wants to kill specific GFP labelled single cells in a zebra fish and follow the developmental consequences. Multiphoton or confocal used to zap a selected area was the suggestion but Is bleaching the GFP sufficient to damage the cell. How can we determine if the cell is dead/damaged - trypan blue in fish. The extent to which surrounding cells are also damaged. Are there any photoactivatable toxins ? Ideas, references, suggestions would be greatly appreciated. Jeremy Adler =============================================== B i o V i s P l a t f o r m of Uppsala University Light & EM microscopy / FlowCytometry & Cell Sorting / Image Analysis =============================================== Jeremy Adler PhD - Senior research engineer Light, Confocal Microscopy, Image Analysis E-mail: [hidden email] 070-1679349 Dag Hammarskjölds v 20 751 85 UPPSALA, SWEDEN http://biovis.uu.se/ =============================================== När du har kontakt med oss på Uppsala universitet med e-post så innebär det att vi behandlar dina personuppgifter. För att läsa mer om hur vi gör det kan du läsa här: http://www.uu.se/om-uu/dataskydd-personuppgifter/ E-mailing Uppsala University means that we will process your personal data. For more information on how this is performed, please read here: http://www.uu.se/en/about-uu/data-protection-policy |
![]() ![]() |
Henrik Persson |
![]() |
In reply to this post by Tim Feinstein
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Jeremy, I have not done this but in this paper I read a while back (Liu et al., Immunity 2016 <https://www.sciencedirect.com/science/article/pii/S1074761316300978>), they used 2-photon microscopy (800 nm) to kill single endothelial cells in zebra fish brain to study blood vessel repair using intravital microscopy. From their methods: *"Laser Microsurgical Cell Ablation, * Photoconversion, In Vivo Time- Lapse, and FRET Imaging For laser microsurgical cell ablation, the target endothelial cell or macrophage was focused in a single confocal plane and irradiated for 3 s by a multi-photon laser at 800 nm (Chameleon Vision2 Laser System, Coherent)." Hope this helps Best wishes /Henrik ______________________ Henrik Persson, Ph.D. Translational Biology & Engineering Program Mechanical and Industrial Engineering University of Toronto, Canada LinkedIn <http://www.linkedin.com/in/henrikcpersson> On Thu, 15 Aug 2019 at 11:07, Feinstein, Timothy N <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Jeremy, > > I'd recommend using a mitochondrial or nuclear-targeted KillerRed for that > purpose. You'd need a pretty modest 555nm light flux to zap targeted cells > into apoptosis. There are probably newer and better genetically targeted > photosensitizing agents, but that's the one that I've used. > > Best, > > > T > > Timothy Feinstein, Ph.D. > Research Scientist > Department of Developmental Biology > University of Pittsburgh > > > > > > On 8/15/19, 7:42 AM, "Confocal Microscopy List on behalf of Jeremy Adler" > <[hidden email] on behalf of [hidden email]> > wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > > https://nam05.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7Ctnf8%40PITT.EDU%7Cf35f8011aca548894d7108d721759a93%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637014661334143515&sdata=pX1oV6q0ojLWNiot%2B6k3cF%2BSa%2B1CIgp6kBFFqD3vDhY%3D&reserved=0 > Post images on > https://nam05.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com&data=02%7C01%7Ctnf8%40PITT.EDU%7Cf35f8011aca548894d7108d721759a93%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637014661334143515&sdata=2oGrBtdIHgZ7WBT04bL9GfVCr9Ce9rQhkNIVgzUPdp4%3D&reserved=0 > and include the link in your posting. > ***** > > > We have a user who wants to kill specific GFP labelled single cells in > a zebra fish and follow the developmental consequences. > Multiphoton or confocal used to zap a selected area was the suggestion > but > > Is bleaching the GFP sufficient to damage the cell. > How can we determine if the cell is dead/damaged - trypan blue in fish. > The extent to which surrounding cells are also damaged. > Are there any photoactivatable toxins ? > > Ideas, references, suggestions would be greatly appreciated. > > > Jeremy Adler > =============================================== > B i o V i s P l a t f o r m of Uppsala > University > Light & EM microscopy / FlowCytometry & Cell > Sorting / Image Analysis =============================================== > Jeremy Adler PhD - Senior research engineer Light, Confocal > Microscopy, Image Analysis > E-mail: [hidden email] > 070-1679349 > > Dag Hammarskjölds v 20 > 751 85 UPPSALA, SWEDEN > > https://nam05.safelinks.protection.outlook.com/?url=http%3A%2F%2Fbiovis.uu.se%2F&data=02%7C01%7Ctnf8%40PITT.EDU%7Cf35f8011aca548894d7108d721759a93%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637014661334143515&sdata=n0bjom4QhdtLqoa0wFb3aaaAfUsgq0t%2FNQa%2BdWZbGA4%3D&reserved=0 > =============================================== > > > > > > > > > > > > > > > > > När du har kontakt med oss på Uppsala universitet med e-post så > innebär det att vi behandlar dina personuppgifter. För att läsa mer om hur > vi gör det kan du läsa här: > https://nam05.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.uu.se%2Fom-uu%2Fdataskydd-personuppgifter%2F&data=02%7C01%7Ctnf8%40PITT.EDU%7Cf35f8011aca548894d7108d721759a93%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637014661334143515&sdata=YPoOYRsJhBDx8rp%2BEYeet3KQjNcU32wsE%2BBF0jOqCXk%3D&reserved=0 > > E-mailing Uppsala University means that we will process your personal > data. For more information on how this is performed, please read here: > https://nam05.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.uu.se%2Fen%2Fabout-uu%2Fdata-protection-policy&data=02%7C01%7Ctnf8%40PITT.EDU%7Cf35f8011aca548894d7108d721759a93%7C9ef9f489e0a04eeb87cc3a526112fd0d%7C1%7C0%7C637014661334143515&sdata=nKa%2FEM0NxEJqxUJNiadUbCSxJGHAq3dDh2djGjIKkAE%3D&reserved=0 > > > |
![]() ![]() |
Advanced Imaging Lab |
![]() |
In reply to this post by Jeremy Adler-4
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Jeremy, You can try using a laser-capture microdissection microscope. We have used it to cut a zebrafish axon and the fish survived the procedure. Best regards, Esther *Dr. Esther G.L. Koh* :: Head, Advanced Imaging Laboratory :: Life Sciences Institute Immunology Programme :: National University of Singapore :: Centre for Life Sciences, 28 Medical Drive #03-06E, Singapore 117456 On Thu, 15 Aug 2019 at 19:42, Jeremy Adler <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > > We have a user who wants to kill specific GFP labelled single cells in a > zebra fish and follow the developmental consequences. > Multiphoton or confocal used to zap a selected area was the suggestion but > > Is bleaching the GFP sufficient to damage the cell. > How can we determine if the cell is dead/damaged - trypan blue in fish. > The extent to which surrounding cells are also damaged. > Are there any photoactivatable toxins ? > > Ideas, references, suggestions would be greatly appreciated. > > > Jeremy Adler > =============================================== > B i o V i s P l a t f o r m of Uppsala University > Light & EM microscopy / FlowCytometry & Cell Sorting / > Image Analysis =============================================== > Jeremy Adler PhD - Senior research engineer Light, Confocal Microscopy, > Image Analysis > E-mail: [hidden email] > 070-1679349 > > Dag Hammarskjölds v 20 > 751 85 UPPSALA, SWEDEN > http://biovis.uu.se/ > =============================================== > > > > > > > > > > > > > > > > > När du har kontakt med oss på Uppsala universitet med e-post så innebär > det att vi behandlar dina personuppgifter. För att läsa mer om hur vi gör > det kan du läsa här: http://www.uu.se/om-uu/dataskydd-personuppgifter/ > > E-mailing Uppsala University means that we will process your personal > data. For more information on how this is performed, please read here: > http://www.uu.se/en/about-uu/data-protection-policy > |
![]() |
In reply to this post by Jeremy Adler-4
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Jeremy, It depends a great deal on how deep the GFP+ cells are into the animal and whether the goal is to kill individual or all GFP+ cells. We had great success using blue light output from a dye laser (Andor Micropoint) in killing hair cells in 2-3 day old larvae: https://www.cell.com/developmental-cell/fulltext/S1534-5807(17)30864-X But these are cells are quite superficial. For broader killing, the ROS generating fluorescent proteins others have mentioned should work well as long as the light exposure required doesn't also cause off-target toxicity. For killing individual cells deep inside of tissue, your only real bet is a femtosecond laser. For a recent implementation of this, see this paper from the Ahrens and Keller labs: https://www.nature.com/articles/s41592-018-0221-x Our experience both with femtosecond and nanosecond lasers for killing single cells has been to use damage accumulation- a long train of low energy pulses enough to bleach a visible spot in the nucleus, but not so much that the entire nucleus or cell bleaches. In our hands this has been quite important for limiting off-target damage. It's important for there to be residual signal so one can actually assess cell death- typically one would expect to see the nucleus either condense or fragment anywhere from 5 minutes to 45 minutes after ablation. Hair cells seem to be cleared within minutes after death so, depending on your timelapse imaging rates, the corpse can simply disappear between timepoints. Acute insults like laser-induced cell killing can also cause significant off-target damage on-axis with the laser, so it's also important to watch cells above and below the target for signs of death or damage as this can potentially confound your experiment. Please feel free to reach out if I can provide any more detailed advice! Best, Pavak On Thu, Aug 15, 2019 at 4:42 AM Jeremy Adler <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > > We have a user who wants to kill specific GFP labelled single cells in a > zebra fish and follow the developmental consequences. > Multiphoton or confocal used to zap a selected area was the suggestion but > > Is bleaching the GFP sufficient to damage the cell. > How can we determine if the cell is dead/damaged - trypan blue in fish. > The extent to which surrounding cells are also damaged. > Are there any photoactivatable toxins ? > > Ideas, references, suggestions would be greatly appreciated. > > > Jeremy Adler > =============================================== > B i o V i s P l a t f o r m of Uppsala University > Light & EM microscopy / FlowCytometry & Cell Sorting / > Image Analysis =============================================== > Jeremy Adler PhD - Senior research engineer Light, Confocal Microscopy, > Image Analysis > E-mail: [hidden email] > 070-1679349 > > Dag Hammarskjölds v 20 > 751 85 UPPSALA, SWEDEN > http://biovis.uu.se/ > =============================================== > > > > > > > > > > > > > > > > > När du har kontakt med oss på Uppsala universitet med e-post så innebär > det att vi behandlar dina personuppgifter. För att läsa mer om hur vi gör > det kan du läsa här: http://www.uu.se/om-uu/dataskydd-personuppgifter/ > > E-mailing Uppsala University means that we will process your personal > data. For more information on how this is performed, please read here: > http://www.uu.se/en/about-uu/data-protection-policy > |