Dear all--

I've just spent an hour+ trying to track down a copy of a translation of
Köhler's 1893 paper on what we now call Köhler illumination.  Murphy and
Davidson's "Fundamentals of Light Microscopy and Digital Imaging" cite
it, and Google Scholar shows it as follows:

Köhler illumination centenary: a collection of papers detailing Köhler
illumination
PJ Evennett - Royal Microscopical Society, 1994

I've checked the RMS site, I've searched our library catalog, I've
searched using Google, and found nothing.  The book or article appears
not to exist, but I expect that it's just fallen through a digital crack.

Can anyone out there help me with a more complete citation or a link?

Thanks--

Martin Wessendorf

--
Martin Wessendorf, Ph.D.                   office: (612) 626-0145
Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991
University of Minnesota             Preferred FAX: (612) 624-8118
6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009
Minneapolis, MN  55455                    e-mail: [hidden email]
Kyle Michael Douglass Kyle Michael Douglass
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: "Köhler illumination centenary"

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Martin,
You want the translation to English, right?

https://micro.magnet.fsu.edu/primer/anatomy/kohleroriginal.html

At the bottom it states:


This translation of Dr. Köhler's original paper, which appeared along with one "woodcut" (Figure 1) in Zeitschrift für wissenschaftl. Mikroskopie, 10: pages 433-440 (1893), was first published in "Köhler Illumination Centenary" by the Royal Microscopical Society in 1994.
Cheers,
Kyle

Dr. Kyle M. Douglass
Post-doctoral Researcher
EPFL - The Laboratory of Experimental Biophysics
http://leb.epfl.ch/
http://kmdouglass.github.io

________________________________________
From: Confocal Microscopy List [[hidden email]] on behalf of Martin Wessendorf [[hidden email]]
Sent: Thursday, August 11, 2016 5:20 PM
To: [hidden email]
Subject: "Köhler illumination centenary"

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Dear all--

I've just spent an hour+ trying to track down a copy of a translation of
Köhler's 1893 paper on what we now call Köhler illumination.  Murphy and
Davidson's "Fundamentals of Light Microscopy and Digital Imaging" cite
it, and Google Scholar shows it as follows:

Köhler illumination centenary: a collection of papers detailing Köhler
illumination
PJ Evennett - Royal Microscopical Society, 1994

I've checked the RMS site, I've searched our library catalog, I've
searched using Google, and found nothing.  The book or article appears
not to exist, but I expect that it's just fallen through a digital crack.

Can anyone out there help me with a more complete citation or a link?

Thanks--

Martin Wessendorf

--
Martin Wessendorf, Ph.D.                   office: (612) 626-0145
Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991
University of Minnesota             Preferred FAX: (612) 624-8118
6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009
Minneapolis, MN  55455                    e-mail: [hidden email]
George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: "Köhler illumination centenary"

In reply to this post by Martin Wessendorf-2
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Martin,

Zeiss sponsored the 1994 Centenary collection (mine print was discarded
many moves ago) - hopefully someone at Zeiss US or Germany HQ still has
it (ex. their archives) and they/RMS can get it scanned to PDF. For that
matter, you could ask the RMS to get you in contact with Peter E.

this is why I look in wikipedia: there is also a J RMS article in 1894

https://en.wikipedia.org/wiki/K%C3%B6hler_illumination

  Köhler, August (1893). "Ein neues Beleuchtungsverfahren für
mikrophotographische Zwecke". Zeitschrift für wissenschaftliche
Mikroskopie und für Mikroskopische Technik. 10 (4): 433–440.

Koehler, August (1894). "New Method of Illimination for
Phomicrographical Purposes". Journal of the Royal Microscopical Society.
14: 261–262.

Which should be in the Microscopy (pages 250-281) at

http://onlinelibrary.wiley.com/doi/10.1111/jmi.1894.14.issue-2/issuetoc


The original was referenced in:

http://rsob.royalsocietypublishing.org/content/5/4/150019

12. KöhlerA.1893Ein neues Beleuchtungsverfahren für mikrophotographische
Zwecke.Z. Wiss. Mikrosk.10,433–440.

 From Animaculum to single molecules: 300 years of the light microscope
Adam J. M. Wollman, Richard Nudd, Erik G. Hedlund, Mark C. Leake
Open Biology. Published 29 April 2015.DOI: 10.1098/rsob.150019


J RMS ... volume 2 was in Internet Archive (I did not see volume 14 -
did not look much)
https://archive.org/details/journalofroyalmi02roya


George


On 8/11/2016 10:20 AM, Martin Wessendorf wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Dear all--
>
> I've just spent an hour+ trying to track down a copy of a translation
> of Köhler's 1893 paper on what we now call Köhler illumination.  
> Murphy and Davidson's "Fundamentals of Light Microscopy and Digital
> Imaging" cite it, and Google Scholar shows it as follows:
>
> Köhler illumination centenary: a collection of papers detailing Köhler
> illumination
> PJ Evennett - Royal Microscopical Society, 1994
>
> I've checked the RMS site, I've searched our library catalog, I've
> searched using Google, and found nothing.  The book or article appears
> not to exist, but I expect that it's just fallen through a digital crack.
>
> Can anyone out there help me with a more complete citation or a link?
>
> Thanks--
>
> Martin Wessendorf
>

--


George McNamara, PhD
Houston, TX 77054
[hidden email]
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/
Michelle Peckham Michelle Peckham
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: "Köhler illumination centenary"

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

By Peter E, do you mean Peter Evenett?

I have contact details for him, if you want to contact me off list.

On 11/08/2016, 16:53, "Confocal Microscopy List on behalf of George McNamara" <[hidden email] on behalf of [hidden email]> wrote:

    *****
    To join, leave or search the confocal microscopy listserv, go to:
    http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    Post images on http://www.imgur.com and include the link in your posting.
    *****
   
    Hi Martin,
   
    Zeiss sponsored the 1994 Centenary collection (mine print was discarded
    many moves ago) - hopefully someone at Zeiss US or Germany HQ still has
    it (ex. their archives) and they/RMS can get it scanned to PDF. For that
    matter, you could ask the RMS to get you in contact with Peter E.
   
    this is why I look in wikipedia: there is also a J RMS article in 1894
   
    https://en.wikipedia.org/wiki/K%C3%B6hler_illumination
   
      Köhler, August (1893). "Ein neues Beleuchtungsverfahren für
    mikrophotographische Zwecke". Zeitschrift für wissenschaftliche
    Mikroskopie und für Mikroskopische Technik. 10 (4): 433–440.
   
    Koehler, August (1894). "New Method of Illimination for
    Phomicrographical Purposes". Journal of the Royal Microscopical Society.
    14: 261–262.
   
    Which should be in the Microscopy (pages 250-281) at
   
    http://onlinelibrary.wiley.com/doi/10.1111/jmi.1894.14.issue-2/issuetoc
   
   
    The original was referenced in:
   
    http://rsob.royalsocietypublishing.org/content/5/4/150019
   
    12. KöhlerA.1893Ein neues Beleuchtungsverfahren für mikrophotographische
    Zwecke.Z. Wiss. Mikrosk.10,433–440.
   
     From Animaculum to single molecules: 300 years of the light microscope
    Adam J. M. Wollman, Richard Nudd, Erik G. Hedlund, Mark C. Leake
    Open Biology. Published 29 April 2015.DOI: 10.1098/rsob.150019
   
   
    J RMS ... volume 2 was in Internet Archive (I did not see volume 14 -
    did not look much)
    https://archive.org/details/journalofroyalmi02roya
   
   
    George
   
   
    On 8/11/2016 10:20 AM, Martin Wessendorf wrote:
    > *****
    > To join, leave or search the confocal microscopy listserv, go to:
    > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    > Post images on http://www.imgur.com and include the link in your posting.
    > *****
    >
    > Dear all--
    >
    > I've just spent an hour+ trying to track down a copy of a translation
    > of Köhler's 1893 paper on what we now call Köhler illumination.  
    > Murphy and Davidson's "Fundamentals of Light Microscopy and Digital
    > Imaging" cite it, and Google Scholar shows it as follows:
    >
    > Köhler illumination centenary: a collection of papers detailing Köhler
    > illumination
    > PJ Evennett - Royal Microscopical Society, 1994
    >
    > I've checked the RMS site, I've searched our library catalog, I've
    > searched using Google, and found nothing.  The book or article appears
    > not to exist, but I expect that it's just fallen through a digital crack.
    >
    > Can anyone out there help me with a more complete citation or a link?
    >
    > Thanks--
    >
    > Martin Wessendorf
    >
   
    --
   
   
    George McNamara, PhD
    Houston, TX 77054
    [hidden email]
    https://www.linkedin.com/in/georgemcnamara
    https://works.bepress.com/gmcnamara/75/
   

Michelle Peckham Michelle Peckham
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: "Köhler illumination centenary"

In reply to this post by Kyle Michael Douglass
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

And: a message from the RMS:

Hi Martin

I have located a copy in the Royal Microscopical Society Office here in Oxford, and I'll contact you (and Peter Evennett) directly.

Best wishes,

Allison

Allison Winton
Event Director
Royal Microscopical Society (RMS)
37/38 St Clements, Oxford  OX4 1AJ, UK
Direct Tel: +44 (0)1865 254764, Mobile: +44 (0)7545 438794 RMS Office: +44 (0)1865 254760, Email [hidden email]  www.rms.org.uk

On 11/08/2016, 16:52, "Confocal Microscopy List on behalf of Kyle Michael Douglass" <[hidden email] on behalf of [hidden email]> wrote:

    *****
    To join, leave or search the confocal microscopy listserv, go to:
    http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    Post images on http://www.imgur.com and include the link in your posting.
    *****
   
    Hi Martin,
    You want the translation to English, right?
   
    https://micro.magnet.fsu.edu/primer/anatomy/kohleroriginal.html
   
    At the bottom it states:
   
   
    This translation of Dr. Köhler's original paper, which appeared along with one "woodcut" (Figure 1) in Zeitschrift für wissenschaftl. Mikroskopie, 10: pages 433-440 (1893), was first published in "Köhler Illumination Centenary" by the Royal Microscopical Society in 1994.
    Cheers,
    Kyle
   
    Dr. Kyle M. Douglass
    Post-doctoral Researcher
    EPFL - The Laboratory of Experimental Biophysics
    http://leb.epfl.ch/
    http://kmdouglass.github.io
   
    ________________________________________
    From: Confocal Microscopy List [[hidden email]] on behalf of Martin Wessendorf [[hidden email]]
    Sent: Thursday, August 11, 2016 5:20 PM
    To: [hidden email]
    Subject: "Köhler illumination centenary"
   
    *****
    To join, leave or search the confocal microscopy listserv, go to:
    http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
    Post images on http://www.imgur.com and include the link in your posting.
    *****
   
    Dear all--
   
    I've just spent an hour+ trying to track down a copy of a translation of
    Köhler's 1893 paper on what we now call Köhler illumination.  Murphy and
    Davidson's "Fundamentals of Light Microscopy and Digital Imaging" cite
    it, and Google Scholar shows it as follows:
   
    Köhler illumination centenary: a collection of papers detailing Köhler
    illumination
    PJ Evennett - Royal Microscopical Society, 1994
   
    I've checked the RMS site, I've searched our library catalog, I've
    searched using Google, and found nothing.  The book or article appears
    not to exist, but I expect that it's just fallen through a digital crack.
   
    Can anyone out there help me with a more complete citation or a link?
   
    Thanks--
   
    Martin Wessendorf
   
    --
    Martin Wessendorf, Ph.D.                   office: (612) 626-0145
    Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991
    University of Minnesota             Preferred FAX: (612) 624-8118
    6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009
    Minneapolis, MN  55455                    e-mail: [hidden email]
   

Martin Wessendorf-2 Martin Wessendorf-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: "Köhler illumination centenary"

In reply to this post by Kyle Michael Douglass
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

THANK YOU to Kyle and everyone else who responded. I hope to have the
students in my microscopy course use the paper, and it looks as if the
text of the translation is available on-line as html, using the link
that Kyle posted.

I appreciate the help!!

Martin



On 8/11/2016 10:52 AM, Kyle Michael Douglass wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Martin,
> You want the translation to English, right?
>
> https://micro.magnet.fsu.edu/primer/anatomy/kohleroriginal.html
>
> At the bottom it states:
>
>
> This translation of Dr. Köhler's original paper, which appeared along with one "woodcut" (Figure 1) in Zeitschrift für wissenschaftl. Mikroskopie, 10: pages 433-440 (1893), was first published in "Köhler Illumination Centenary" by the Royal Microscopical Society in 1994.
> Cheers,
> Kyle
>
> Dr. Kyle M. Douglass
> Post-doctoral Researcher
> EPFL - The Laboratory of Experimental Biophysics
> http://leb.epfl.ch/
> http://kmdouglass.github.io
>
> ________________________________________
> From: Confocal Microscopy List [[hidden email]] on behalf of Martin Wessendorf [[hidden email]]
> Sent: Thursday, August 11, 2016 5:20 PM
> To: [hidden email]
> Subject: "Köhler illumination centenary"
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Dear all--
>
> I've just spent an hour+ trying to track down a copy of a translation of
> Köhler's 1893 paper on what we now call Köhler illumination.  Murphy and
> Davidson's "Fundamentals of Light Microscopy and Digital Imaging" cite
> it, and Google Scholar shows it as follows:
>
> Köhler illumination centenary: a collection of papers detailing Köhler
> illumination
> PJ Evennett - Royal Microscopical Society, 1994
>
> I've checked the RMS site, I've searched our library catalog, I've
> searched using Google, and found nothing.  The book or article appears
> not to exist, but I expect that it's just fallen through a digital crack.
>
> Can anyone out there help me with a more complete citation or a link?
>
> Thanks--
>
> Martin Wessendorf
>
> --
> Martin Wessendorf, Ph.D.                   office: (612) 626-0145
> Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991
> University of Minnesota             Preferred FAX: (612) 624-8118
> 6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009
> Minneapolis, MN  55455                    e-mail: [hidden email]

--
Martin Wessendorf, Ph.D.                   office: (612) 626-0145
Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991
University of Minnesota             Preferred FAX: (612) 624-8118
6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009
Minneapolis, MN  55455                    e-mail: [hidden email]
Benjamin Smith Benjamin Smith
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: "Köhler illumination centenary"

In reply to this post by Martin Wessendorf-2
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I've tried to track down articles from that era on Google books with pretty
good success.  Since most of the scanning is automated, sometimes there can
be a mistake in the conversion from print text to digital text.  Therefore,
seeing if you can find that volume in Google books, or other volumes close
to it in time will help narrow down how the find the article you are
looking for.  Unfortunately, sometimes it requires guessing what the
accidental autocorrect, which can be fairly challenging for those of us who
are not fluent in German.

Once you find it, Google translate does a spectacular job at translating,
since there are a LOT of cognates between German and English, especially
when you get into scientific jargon.

Good luck,
   Ben Smith

On Thu, Aug 11, 2016 at 8:20 AM, Martin Wessendorf <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Dear all--
>
> I've just spent an hour+ trying to track down a copy of a translation of
> Köhler's 1893 paper on what we now call Köhler illumination.  Murphy and
> Davidson's "Fundamentals of Light Microscopy and Digital Imaging" cite it,
> and Google Scholar shows it as follows:
>
> Köhler illumination centenary: a collection of papers detailing Köhler
> illumination
> PJ Evennett - Royal Microscopical Society, 1994
>
> I've checked the RMS site, I've searched our library catalog, I've
> searched using Google, and found nothing.  The book or article appears not
> to exist, but I expect that it's just fallen through a digital crack.
>
> Can anyone out there help me with a more complete citation or a link?
>
> Thanks--
>
> Martin Wessendorf
>
> --
> Martin Wessendorf, Ph.D.                   office: (612) 626-0145
> Assoc Prof, Dept Neuroscience                 lab: (612) 624-2991
> University of Minnesota             Preferred FAX: (612) 624-8118
> 6-145 Jackson Hall, 321 Church St. SE    Dept Fax: (612) 626-5009
> Minneapolis, MN  55455                    e-mail: [hidden email]
>



--
Benjamin E. Smith, Ph. D.
Microscopy Specialist, Vision Science
University of California, Berkeley
142 Life Sciences Addition
Berkeley, CA  94720-3200
Tel  (510) 642-9715
Fax (510) 643-6791
e-mail: [hidden email]
http://vision.berkeley.edu/