Ekaterina PAPUSHEVA-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi All, Does anyone knows a light heating insert that can be used with Z-galvo stage from Leica? Would be grateful for a tip! Best regards, Ekaterina ___ Ekaterina Papusheva, PhD Imaging Facility Manager Institute of Science and Technology Austria Phone +43-(0)2243 9000-1043; Mobile +43-(0)664 88509130 Fax +43-(0)2243 9000-2000 Am Campus 1 A-3400 Klosterneuburg Visit our website: www.ist.ac.at<http://www.ist.ac.at> This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. |
|
Jacqueline Ross |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Ektarina, We have a stage adapter that we bought from Warner Instruments, which fits their Chamber Quick Change Heated Base (QE-1). The adapter is called a "Series 20 Stage Adapter for Leica Galvo Z Stage (SA-20GALVOZ)". We use it on our Leica TCS SP2, which has an upright stand (DMRXA2). Kind regards, Jacqui Jacqueline Ross Biomedical Imaging Microscopist Biomedical Imaging Research Unit School of Medical Sciences Faculty of Medical & Health Sciences The University of Auckland Private Bag 92019 Auckland, NEW ZEALAND Tel: 64 9 373 7599 Ext 87438 Fax: 64 9 373 7484 http://www.fmhs.auckland.ac.nz/sms/biru/ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ekaterina PAPUSHEVA Sent: Monday, 4 July 2011 10:43 p.m. To: [hidden email] Subject: Leica Z-galvo heating insert ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi All, Does anyone knows a light heating insert that can be used with Z-galvo stage from Leica? Would be grateful for a tip! Best regards, Ekaterina ___ Ekaterina Papusheva, PhD Imaging Facility Manager Institute of Science and Technology Austria Phone +43-(0)2243 9000-1043; Mobile +43-(0)664 88509130 Fax +43-(0)2243 9000-2000 Am Campus 1 A-3400 Klosterneuburg Visit our website: www.ist.ac.at<http://www.ist.ac.at> This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. |
|
Stephen Firth (Med) |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Ekaterina, we made an aluminium ring which fits the Leica galvo stage and wrapped this with resistors. When a voltage is applied the resistors heat up, these can be easily calibrated for the voltage applied vs temperature. I don't know if proper plans were ever made but I can inquire if you or the list is interested. Cheers, Stephen -- Stephen Firth Manager of Advanced Optical Microscopy Monash Micro Imaging Monash University Room G58 Building 75 Phone 990 55612 Mob 0400 695 425 |
|
|
In reply to this post by Ekaterina PAPUSHEVA-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Ekaterina BIoptechs makes numerous micro-environmental systems to fit Galvanic stages. Leica has several models of Z Galvo stages. Hereis a link to some of the possibilities of the integration of a Delta T™ system. http://www.bioptechs.com/Products/Delta_T/DStages/dstages.html (Scroll down to the bottom of the page) There is also our Stable Z System that accommodates standard dishes. It supplies peripheral heat just as any other "stage heater" but it does not induce Z axis drift. If you are using a Galvo stage you are probably already working with narrow depth of field objectives. Therefore, please beware of simple heating plate designs or as they are traditionally known "stage heaters" because they typically induce thermal drift. http://www.bioptechs.com/Products/Stable_Z/Stable_Z.html Bioptechs specializes in accurately identifying your exact needs and matching the most appropriate hardware. Here is a link to our user profile Questionnaire. The answers to these questions tell us what kind of micro-environmental system will best suit your needs. http://www.bioptechs.com/ProfileQuestions.pdf If you are interested, we can connect you with a distributor that services Austria Dan On Jul 4, 2011, at 6:42 AM, Ekaterina PAPUSHEVA wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi All, Does anyone knows a light heating insert that can be used with Z-galvo stage from Leica? Would be grateful for a tip! Best regards, Ekaterina ___ Ekaterina Papusheva, PhD Imaging Facility Manager Institute of Science and Technology Austria Phone +43-(0)2243 9000-1043; Mobile +43-(0)664 88509130 Fax +43-(0)2243 9000-2000 Am Campus 1 A-3400 Klosterneuburg Visit our website: www.ist.ac.at<http://www.ist.ac.at> This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. Dan Focht Bioptechs, Inc. 3560 Beck Rd. Butler, PA 16002 www.bioptechs.com P: (724)282-7145 F: (724)282-0745 [hidden email] |
|
Ekaterina PAPUSHEVA-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Dan, Would be great if we could know more about your products. Our Z-galvo stage consists of following Leica products: SuperZ Galvo Stage Type H SuperZ Galvo-Tisch Reproduzierbarkeit 40nm Verfahrbereich in Z 500 mm No : 156504116 200 Insert universal Type H Universaleinsatz für SuperZ-Tisch, Typ H Für DMI 6000 mit SuperZ-Tisch No : 156504115 We mainly use 5 cm glass-bottom dishes from MaTek corporation, Cat. Nr. P50G-0-14-F Would be great if you could inform us about the possibilities, and of course considering Z-drift is very important! Best regards, Ekaterina ___ Ekaterina Papusheva, PhD Imaging Facility Manager Institute of Science and Technology Austria Phone +43-(0)2243 9000-1043; Mobile +43-(0)664 88509130 Fax +43-(0)2243 9000-2000 Am Campus 1 A-3400 Klosterneuburg Visit our website: www.ist.ac.at This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Dan Focht Sent: Tuesday, July 05, 2011 4:23 PM To: [hidden email] Subject: Re: Leica Z-galvo heating insert Commercial reply ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Ekaterina BIoptechs makes numerous micro-environmental systems to fit Galvanic stages. Leica has several models of Z Galvo stages. Hereis a link to some of the possibilities of the integration of a Delta TT system. http://www.bioptechs.com/Products/Delta_T/DStages/dstages.html (Scroll down to the bottom of the page) There is also our Stable Z System that accommodates standard dishes. It supplies peripheral heat just as any other "stage heater" but it does not induce Z axis drift. If you are using a Galvo stage you are probably already working with narrow depth of field objectives. Therefore, please beware of simple heating plate designs or as they are traditionally known "stage heaters" because they typically induce thermal drift. http://www.bioptechs.com/Products/Stable_Z/Stable_Z.html Bioptechs specializes in accurately identifying your exact needs and matching the most appropriate hardware. Here is a link to our user profile Questionnaire. The answers to these questions tell us what kind of micro-environmental system will best suit your needs. http://www.bioptechs.com/ProfileQuestions.pdf If you are interested, we can connect you with a distributor that services Austria Dan On Jul 4, 2011, at 6:42 AM, Ekaterina PAPUSHEVA wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi All, Does anyone knows a light heating insert that can be used with Z-galvo stage from Leica? Would be grateful for a tip! Best regards, Ekaterina ___ Ekaterina Papusheva, PhD Imaging Facility Manager Institute of Science and Technology Austria Phone +43-(0)2243 9000-1043; Mobile +43-(0)664 88509130 Fax +43-(0)2243 9000-2000 Am Campus 1 A-3400 Klosterneuburg Visit our website: www.ist.ac.at<http://www.ist.ac.at> This message and any attachment (together "the Message") are confidential and intended only for the individual named. If you have received the Message in error you should not distribute or copy this Message. Please notify the sender immediately by e-mail if you have received this Message in error and delete the Message from your system. Correspondence via e-mail is for information purposes. IST Austria neither makes nor accepts legally binding statements unless explicitly otherwise agreed to the contrary. Dan Focht Bioptechs, Inc. 3560 Beck Rd. Butler, PA 16002 www.bioptechs.com P: (724)282-7145 F: (724)282-0745 [hidden email] |