![]() ![]() |
Kost, Benedikt |
![]() |
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Can anyone compare the performance of
the Leica SP5 resonant scanner with
that of a spinning disk (yokagawa) confocal system in terms of - image quality - photobleaching/phototxicity - general "usefulness" for live cell confocal imaging The resonant scanner was recently demonstrated to us and I was quite impressed. I guess it is not as fast as spinning disk confocals can be, but you seem to be able to image at a rate of at least 10 fps at 512x512, which is all we need. The images we got were very nice, but I did not have the chance to do extensive tests. I would appreciate very much comments
from anyone who has experience using the resonant scanner for live
cell confocal imaging.
Many thanks, Benedikt Kost ____________________________________________________ Benedikt Kost, PhD Warwick HRI Wellesbourne Warwickshire, CV35 9EF UK Tel: +44 (0)24 7657 5092 Fax: +44 (0)24 7657 4500 e-mail: b.kost@... internet: http://www2.warwick.ac.uk/fac/sci/whri/ |
![]() ![]() |
Friedrich Johenning |
![]() |
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear Benedikt, I used a custom-built (not the Leica-Product) resonant scanner and a spinning-disc confocal with a fast CCD camera for calcium imaging in dendrites and spines of patched neurons in acute brain slices. Although we never systematically quantified signal to noise ratio vs. phototoxicity, from my own experience it seems that the spinning disc confocal is better when it comes to photodamage. The reason is that resonant scanners have a very short pixel dwell time, which you have to compensate for by illumination intensity, whereas the spinning disc uses a parallel scan which permits for longer exposures repeated several times during one illumination frame. Wang et al., Journal of Microscopy, 2005 is a very nice study where the difference between parallel scanning and single-point scanning is thoroughly analysed. When dealing with large fluorescent volumes, spinning disc confocals have the problem of crosstalk between adjacent channels, but this dependes on the type of preparation you use. A disadvantage of the spinning disc is that it only has one pinhole size which limits the objectives you can use. Maybe you should also look into the visitec infinity 3 scanner if you want parallel scanning with different pinhole sizes. Best, Friedrich Dr. Friedrich W. Johenning NWFZ -Campus Mitte / AG Schmitz Charite der Universitätsmedizin Berlin Charitéplatz 1 10117 Berlin Tel.: 030 450 528077 Mobil: 0173 2144701 Fax: 030 450 539943 |
![]() ![]() |
Rietdorf, Jens |
![]() |
In reply to this post by Kost, Benedikt
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Dear Benedikt,
| in my opinion,
the major advantage of a parallel compared to single beam scanner is a result
of the low saturation level of the fluorophores (the laser power is split
up to many beamlets or a line or other pattern) and therefore reducing the
population of molecules in dark states, while keeping a high overall photon
flux. In resonant scanners, this beneficial effect (also) results
from avoiding dark states, its descibed in this
article:
Donnert, G., C. Eggeling and
S. W. Hell: Major signal increase in fluorescence
microscopy through dark-state relaxation. | Nature Methods
4 81-86 (2007) | Among
the parallel scanning devices, the spinning disk is (again in my opinion) the
most efficient device, because it avoids turning times of scanners, which are
dead times for the acquisition, because beams will have to be blanked during the
turnaround. Parallel scanners with CCD detectors typically outperform PMTs in terms
of QE over a broad wavelength range.
Point
scanners on the other hand are more flexible, ie you can scan lines or
AOIs, rotate scanfields etc., also confocality is not compromised under any
circumstance or in any direction. For deep-in tissue imaging (above 50microm),
single beam scanners typically outperform parallel scanners.
For
full frame confocal of living specimen of moderate thickness (up to 25microm),
my favorite scanner is still the Yokogawa CSU-22 (no commercial interest) in
combinaton with a EMCCD camera and powerful lasers >25mW.
regards, jens
--- Dr.
Jens Rietdorf[hidden email] From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Kost, Benedikt Sent: Montag, 12. November 2007 14:03 To: [hidden email] Subject: Leica resonant scanner-live for cell imaging Can anyone compare the performance of
the Leica SP5 resonant scanner with
that of a spinning disk (yokagawa) confocal system in terms of - image quality - photobleaching/phototxicity - general "usefulness" for live cell confocal imaging The resonant scanner was recently demonstrated to us and I was quite impressed. I guess it is not as fast as spinning disk confocals can be, but you seem to be able to image at a rate of at least 10 fps at 512x512, which is all we need. The images we got were very nice, but I did not have the chance to do extensive tests. I would appreciate very much comments from anyone who has experience
using the resonant scanner for live cell confocal imaging.
Many thanks, Benedikt Kost ____________________________________________________ Benedikt Kost, PhD Warwick HRI Wellesbourne Warwickshire, CV35 9EF UK Tel: +44 (0)24 7657 5092 Fax: +44 (0)24 7657 4500 e-mail: b.kost@... internet: http://www2.warwick.ac.uk/fac/sci/whri/ |
![]() ![]() |
Martin Seem |
![]() |
In reply to this post by Kost, Benedikt
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Benedikt, we acquired a Leica resonant scanner based confocal about 6 months ago. Before buying this system we tested different types of microscopes including three spinning disc based systems (Andor, PE and Visitec QLC100), two point-scan based systems (Leica and Visitec Eye) and the new Zeiss LSM5 Live system. We brought our own samples (transgenic plant material) and used these to test the different systems in terms of speed, photobleaching and overall image quality. We found the spinning disc based systems to be slightly faster than the Leica system especially when acquiring square images (e.g. 512x512 pixels). However the speed of the Leica system increases significantly when you reduce the number of lines in the Y direction. You should have no problem acquiring 10 fps with a resolution of 512x512 pixels resolution and about 15-20 fps with a resolution of 512x128 pixels. In our experience the level of photobleaching is quite low when using the Leica system. I would say it is comparable to the spinning disc systems we tested and significantly lower than the Visitec Eye system that bleached my sample (transgenic GFP5-ER plants) within seconds (this system did however work nicely for calcium imaging in dendrites so the level of photobleaching depends on the properties of the fluorechrome). In the end we chose the Leica system because it was sufficiently fast, had an adjustable and round pinhole making it fully confocal, had low levels of photobleaching, had a very flexible bandpass filter solution and could easily and relatively cheaply be upgraded to a tandem scanner system (both resonant and conventional scanner in the same system). Best regards Martin Seem ------------------------------------------------------------------------- Martin Seem Graduate student Group for Cell and Molecular Biology Norwegian University of Science and Technology |
![]() ![]() |
Ken Bell-2 |
![]() |
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear Gentlemen, My name is Lukasz Kilianek. I'm a Product Specialist at VisiTech International -manufacturer of VT Infinity and VT Eye confocal units. I follow your discussion about system comparison in term of speed and photobleaching. Let me point out that original mail was about VT Infinity. The last mail pertains to VT Eye system. Let me explain here - VT Infinity is a multiple point scanner, VY Eye is a single spot scanner. It makes a huge difference in performance. With best regards Lukasz Kilianek Product Specialist VisiTech International Ltd Unit 92, Silverbriar Sunderland Enterprise Park (East) Sunderland SR5 2TQ, UK Tel +44 191 516 6255 Fax +44 191 516 6258 email: [hidden email] -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Martin Seem Sent: 13 November 2007 13:57 To: [hidden email] Subject: Re: Leica resonant scanner-live for cell imaging Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Benedikt, we acquired a Leica resonant scanner based confocal about 6 months ago. Before buying this system we tested different types of microscopes including three spinning disc based systems (Andor, PE and Visitec QLC100), two point-scan based systems (Leica and Visitec Eye) and the new Zeiss LSM5 Live system. We brought our own samples (transgenic plant material) and used these to test the different systems in terms of speed, photobleaching and overall image quality. We found the spinning disc based systems to be slightly faster than the Leica system especially when acquiring square images (e.g. 512x512 pixels). However the speed of the Leica system increases significantly when you reduce the number of lines in the Y direction. You should have no problem acquiring 10 fps with a resolution of 512x512 pixels resolution and about 15-20 fps with a resolution of 512x128 pixels. In our experience the level of photobleaching is quite low when using the Leica system. I would say it is comparable to the spinning disc systems we tested and significantly lower than the Visitec Eye system that bleached my sample (transgenic GFP5-ER plants) within seconds (this system did however work nicely for calcium imaging in dendrites so the level of photobleaching depends on the properties of the fluorechrome). In the end we chose the Leica system because it was sufficiently fast, had an adjustable and round pinhole making it fully confocal, had low levels of photobleaching, had a very flexible bandpass filter solution and could easily and relatively cheaply be upgraded to a tandem scanner system (both resonant and conventional scanner in the same system). Best regards Martin Seem ------------------------------------------------------------------------ - Martin Seem Graduate student Group for Cell and Molecular Biology Norwegian University of Science and Technology |
![]() ![]() |
James Pawley |
![]() |
In reply to this post by Martin Seem
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >Search the CONFOCAL archive at >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >Hi Benedikt, > >we acquired a Leica resonant scanner based confocal about 6 months >ago. Before buying this system we tested different types of >microscopes including three spinning disc based systems (Andor, PE >and Visitec QLC100), two point-scan based systems (Leica and Visitec >Eye) and the new Zeiss LSM5 Live system. We brought our own samples >(transgenic plant material) and used these to test the different >systems in terms of speed, photobleaching and overall image quality. > >We found the spinning disc based systems to be slightly faster than >the Leica system especially when acquiring square images (e.g. >512x512 pixels). However the speed of the Leica system increases >significantly when you reduce the number of lines in the Y >direction. You should have no problem acquiring 10 fps with a >resolution of 512x512 pixels resolution and about 15-20 fps with a >resolution of 512x128 pixels. > >In our experience the level of photobleaching is quite low when >using the Leica system. I would say it is comparable to the spinning >disc systems we tested and significantly lower than the Visitec Eye >system that bleached my sample (transgenic GFP5-ER plants) within >seconds (this system did however work nicely for calcium imaging in >dendrites so the level of photobleaching depends on the properties >of the fluorechrome). > >In the end we chose the Leica system because it was sufficiently >fast, had an adjustable and round pinhole making it fully confocal, >had low levels of photobleaching, had a very flexible bandpass >filter solution and could easily and relatively cheaply be upgraded >to a tandem scanner system (both resonant and conventional scanner >in the same system). > > >Best regards >Martin Seem >------------------------------------------------------------------------- > >Martin Seem >Graduate student >Group for Cell and Molecular Biology >Norwegian University of Science and Technology Thank you for this Martin, Could you please tell us about the type of CCD cameras used with the various disk-scanners? EM-CCD or normal CCD? Thanks, Jim P. -- **************************************** Prof. James B. Pawley, Ph. 608-263-3147 Room 223, Zoology Research Building, FAX 608-262-9083 250 N. Mills St., Madison, WI, 53706 [hidden email] "A scientist is not one who can answer questions but one who can question answers." Theodore Schick Jr., Skeptical Enquirer, 21-2:39 |
![]() ![]() |
Martin Seem |
![]() |
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Jim, we tested the PE Ultraview at to separate occasions. Once with a normal ORCA CCD-camera at the Molecular Imaging Center (MIC) in Bergen (Norway) and once with an EM CCD camera at PEs training facility at Seer Green right outside London. The VisiTech QLC100 was a CSU10 based model equipped with an ORCA CCD-camera while the Andor system used an EM CCD camera. Thanks, Martin ------------------------------------------------------------------------ Martin Seem Graduate student Group for Cell and Molecular Biology Norwegian University of Science and Technology Quoting James Pawley <[hidden email]>: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > >Search the CONFOCAL archive at > >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > > >Hi Benedikt, > > > >we acquired a Leica resonant scanner based confocal about 6 months > >ago. Before buying this system we tested different types of > >microscopes including three spinning disc based systems (Andor, PE > >and Visitec QLC100), two point-scan based systems (Leica and Visitec > >Eye) and the new Zeiss LSM5 Live system. We brought our own samples > >(transgenic plant material) and used these to test the different > >systems in terms of speed, photobleaching and overall image quality. > > > >We found the spinning disc based systems to be slightly faster than > >the Leica system especially when acquiring square images (e.g. > >512x512 pixels). However the speed of the Leica system increases > >significantly when you reduce the number of lines in the Y > >direction. You should have no problem acquiring 10 fps with a > >resolution of 512x512 pixels resolution and about 15-20 fps with a > >resolution of 512x128 pixels. > > > >In our experience the level of photobleaching is quite low when > >using the Leica system. I would say it is comparable to the spinning > >disc systems we tested and significantly lower than the Visitec Eye > >system that bleached my sample (transgenic GFP5-ER plants) within > >seconds (this system did however work nicely for calcium imaging in > >dendrites so the level of photobleaching depends on the properties > >of the fluorechrome). > > > >In the end we chose the Leica system because it was sufficiently > >fast, had an adjustable and round pinhole making it fully confocal, > >had low levels of photobleaching, had a very flexible bandpass > >filter solution and could easily and relatively cheaply be upgraded > >to a tandem scanner system (both resonant and conventional scanner > >in the same system). > > > > > >Best regards > >Martin Seem > >------------------------------------------------------------------------- > > > >Martin Seem > >Graduate student > >Group for Cell and Molecular Biology > >Norwegian University of Science and Technology > > Thank you for this Martin, > > Could you please tell us about the type of CCD cameras used with the > various disk-scanners? EM-CCD or normal CCD? > > Thanks, > > Jim P. > -- > **************************************** > Prof. James B. Pawley, Ph. 608-263-3147 > Room 223, Zoology Research Building, FAX > 608-262-9083 > 250 N. Mills St., Madison, WI, 53706 [hidden email] > "A scientist is not one who can answer questions but one who can > question answers." Theodore Schick Jr., Skeptical Enquirer, 21-2:39 > |
![]() ![]() |
Stefan Terjung |
![]() |
In reply to this post by James Pawley
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear Benedikt and others, We have a Leica SP5 tandem scanner and a PerkinElmer Ultraview ERS (6 laserlines, Hamamatsu EM CCD with 8µm pixel size). Regarding your questions: - It is easier to setup the fast time-lapse experiment on a spining disk for fast time-lapse imaging (I can only judge for the PerkinElmer). So the spinning disks are frequently used for fast time-lapse imaging in our facility. - But you can get quite comparable time-lapse sequences with a bit more 'fine-tuning' on the SP5 (at least for some cases): a) use accumulation (unfortunately only frame accumulation available yet) to compensate for lower brightness due to shorter pixel dwell time instead of increasing the laser power => this is similar to increasing the exposure time on a spinning disk without saturating the fluorophores b) use less lines (rectangle instead of square format) if possible c) you can acquire up to 3 channels simultaneous or line-by-line sequential leading to the same timing for up to 3 channels compared to frame-by-frame sequential (on the Ultraview - there are also spinning disks available with 2 cameras and simultaneous acquisition) d) often you can open the pinhole at least a bit in time-lapse imaging leading to brighter images (not possible on the Yokogawa spinning disks) e) you can activate the bidirectional (and optimize the phase with the transmission channel as this always has enough brightness and suitable structures) f) you are more flexible adjusting the desired detection range (zoom! Illuminating only the detected region) g) the spinning disk tends to have more problems with thicker samples (crosstalk between the pinholes) - The SP5 (in my opinion) is in a comparable range regarding photobleaching/ phototoxicity as long as one doesn´t use too much laser power (using rather accumulations instead of increasing laser power - see 'a)' above). No commercial interest in both systems. Hope this helps, Regards, Stefan > -----Ursprüngliche Nachricht----- > Von: Confocal Microscopy List > [mailto:[hidden email]] Im Auftrag von James Pawley > Gesendet: Dienstag, 13. November 2007 19:29 > An: [hidden email] > Betreff: Re: Leica resonant scanner-live for cell imaging > > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > >Search the CONFOCAL archive at > >http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > > >Hi Benedikt, > > > >we acquired a Leica resonant scanner based confocal about 6 > months ago. > >Before buying this system we tested different types of microscopes > >including three spinning disc based systems (Andor, PE and Visitec > >QLC100), two point-scan based systems (Leica and Visitec > >Eye) and the new Zeiss LSM5 Live system. We brought our own samples > >(transgenic plant material) and used these to test the different > >systems in terms of speed, photobleaching and overall image quality. > > > >We found the spinning disc based systems to be slightly > faster than the > >Leica system especially when acquiring square images (e.g. > >512x512 pixels). However the speed of the Leica system increases > >significantly when you reduce the number of lines in the Y > direction. > >You should have no problem acquiring 10 fps with a resolution of > >512x512 pixels resolution and about 15-20 fps with a resolution of > >512x128 pixels. > > > >In our experience the level of photobleaching is quite low > when using > >the Leica system. I would say it is comparable to the spinning disc > >systems we tested and significantly lower than the Visitec > Eye system > >that bleached my sample (transgenic GFP5-ER plants) within seconds > >(this system did however work nicely for calcium imaging in > dendrites > >so the level of photobleaching depends on the properties of the > >fluorechrome). > > > >In the end we chose the Leica system because it was > sufficiently fast, > >had an adjustable and round pinhole making it fully > confocal, had low > >levels of photobleaching, had a very flexible bandpass > filter solution > >and could easily and relatively cheaply be upgraded to a > tandem scanner > >system (both resonant and conventional scanner in the same system). > > > > > >Best regards > >Martin Seem > >------------------------------------------------------------- > ------------ > > > >Martin Seem > >Graduate student > >Group for Cell and Molecular Biology > >Norwegian University of Science and Technology > > Thank you for this Martin, > > Could you please tell us about the type of CCD cameras used with the > various disk-scanners? EM-CCD or normal CCD? > > Thanks, > > Jim P. > -- > **************************************** > Prof. James B. Pawley, Ph. 608-263-3147 > Room 223, Zoology Research Building, > FAX 608-262-9083 > 250 N. Mills St., Madison, WI, 53706 [hidden email] > "A scientist is not one who can answer questions but one who can > question answers." Theodore Schick Jr., Skeptical Enquirer, 21-2:39 |
![]() ![]() |
Michael Weber-4 |
![]() |
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Just to add that, line accumulation is available since LAS AF v1.8. So you can finally set it up using "Live" Scan. How does 2x Accumulation compete compared to 2x more laser power? The total illumination power added to one pixel is in the end simliar. So signal-to-noise and bleaching effects should be comparable as well, or not? cheers, Michael Stefan Terjung wrote: > - But you can get quite comparable time-lapse sequences with a bit > more 'fine-tuning' on the SP5 (at least for some cases): > a) use accumulation (unfortunately only frame accumulation > available yet) to compensate for lower brightness due to > shorter pixel dwell time instead of increasing the laser power > => this is similar to increasing the exposure time on a > spinning disk without saturating the fluorophores |
![]() ![]() |
Stefan Terjung |
![]() |
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear Michael, Ok, I should have installed 1.8 obviously, good to know. If you plot fluorescence intensity against excitation intensity you get a saturation curve. As one typically uses high power for point scanning one easily gets into this saturation range. This means that increasing the excitation light even more to compensate for faster scanning you don´t gain as much fluorescence intensity (there is a limited number of fluorophores and if (nearly) all are excited you can´t excite more) and it´s even more likely that you bleach stronger as it´s more likely that you excite a excited fluorophore (which significantly increases the chance to bleach). Regards, Stefan > -----Ursprüngliche Nachricht----- > Von: Confocal Microscopy List > [mailto:[hidden email]] Im Auftrag von Michael Weber > Gesendet: Mittwoch, 28. November 2007 13:45 > An: [hidden email] > Betreff: Re: AW: Leica resonant scanner-live for cell imaging > > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Just to add that, line accumulation is available since LAS AF > v1.8. So you can finally set it up using "Live" Scan. > > How does 2x Accumulation compete compared to 2x more laser > power? The total illumination power added to one pixel is in > the end simliar. So signal-to-noise and bleaching effects > should be comparable as well, or not? > > cheers, > Michael > > > Stefan Terjung wrote: > > - But you can get quite comparable time-lapse sequences with a bit > > more 'fine-tuning' on the SP5 (at least for some cases): > > a) use accumulation (unfortunately only frame accumulation > > available yet) to compensate for lower brightness due to > > shorter pixel dwell time instead of increasing the > laser power > > => this is similar to increasing the exposure time on a > > spinning disk without saturating the fluorophores |
Free forum by Nabble | Edit this page |