Knecht, David |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I have limited experience with multi-photon confocal microscopes. I tried imaging live cells with one a while back and was unimpressed with the results as compared to a standard laser scanning instrument. If you had a new single photon instrument and a new multi-photon instrument sitting side by side in your facility (and I presume some of you do), which would you choose for imaging fixed cells or live cells in culture (not trying to go deep into tissue)? Thanks- Dave Dr. David Knecht Department of Molecular and Cell Biology Co-head Flow Cytometry and Confocal Microscopy Facility U-3125 91 N. Eagleville Rd. University of Connecticut Storrs, CT 06269 860-486-2200 860-486-4331 (fax) |
|
Vladimir Ghukasyan-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi David, For this kind of application MP has no advantage other than no out-of-focus photobleaching. CLSM is a better solution. With regards, Vladimir Ghukasyan Confocal and Multiphoton Imaging Facility Neuroscience Center University of North Carolina 115 Mason Farm Rd., Chapel Hill 27599-7250 Tel.: (919) 966 5807 On Fri, Apr 22, 2011 at 1:29 PM, David Knecht charter <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I have limited experience with multi-photon confocal microscopes. I tried imaging live cells with one a while back and was unimpressed with the results as compared to a standard laser scanning instrument. If you had a new single photon instrument and a new multi-photon instrument sitting side by side in your facility (and I presume some of you do), which would you choose for imaging fixed cells or live cells in culture (not trying to go deep into tissue)? Thanks- Dave > > Dr. David Knecht > Department of Molecular and Cell Biology > Co-head Flow Cytometry and Confocal Microscopy Facility > U-3125 > 91 N. Eagleville Rd. > University of Connecticut > Storrs, CT 06269 > 860-486-2200 > 860-486-4331 (fax) > |
|
Armstrong, Brian |
|
|
In reply to this post by Knecht, David
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** David, I think from a theoretical viewpoint the multiphoton would have worse resolution because of the higher lambda (d = 900nm * 0.61 / NA), but better resolution due to smaller focal spot of 2P, so the resolution should be roughly equal to CLSM. In practice (in my hands at least) CLSM images appear to have much better resolution. I would choose CLSM over 2P for the application you describe. Part of the problem as I see it comes from the wide excitation band of 2PE, meaning that you are exciting many fluorophores with a single wavelength, and you do not have this problem when using a "vis" laser. Also, I think per unit area, in the focal spot, the 2PE will kill live cells faster than 1PE. Cheers, Brian D Armstrong PhD Assistant Research Professor Light Microscopy Core Beckman Research Institute City of Hope Dept of Neuroscience 1450 E Duarte Rd Duarte, CA 91010 626-256-4673 x62872 http://www.cityofhope.org/research/support/Light-Microscopy-Digital-Imaging/Pages/default.aspx -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of David Knecht charter Sent: Friday, April 22, 2011 10:29 AM To: [hidden email] Subject: Live cell imaging with multiphoton confocal ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I have limited experience with multi-photon confocal microscopes. I tried imaging live cells with one a while back and was unimpressed with the results as compared to a standard laser scanning instrument. If you had a new single photon instrument and a new multi-photon instrument sitting side by side in your facility (and I presume some of you do), which would you choose for imaging fixed cells or live cells in culture (not trying to go deep into tissue)? Thanks- Dave Dr. David Knecht Department of Molecular and Cell Biology Co-head Flow Cytometry and Confocal Microscopy Facility U-3125 91 N. Eagleville Rd. University of Connecticut Storrs, CT 06269 860-486-2200 860-486-4331 (fax) --------------------------------------------------------------------- SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. --------------------------------------------------------------------- |
|
Kate Luby-Phelps |
|
|
In reply to this post by Knecht, David
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Multiphoton has no clear advantage for cells in culture. For cell culture samples, we use two photon only to excite DAPI or other UV and near-UV fluors since we had to make a choice between the Ti:S and a 405 laser on our scope. Kate Luby-Phelps |
|
Lemasters, John J. |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Multiphoton microscopy is helpful to excite UV-absorbing fluorophores like NAD(P)H that cannot otherwise be excited without specialized lasers and optics. -- John J. Lemasters, MD, PhD Professor and GlaxoSmithKline Distinguished Endowed Chair Director, Center for Cell Death, Injury and Regeneration Departments of Pharmaceutical & Biomedical Sciences and Biochemistry & Molecular Biology Medical University of South Carolina QF213 Quadrangle Building 280 Calhoun Street, MSC 140 Charleston, SC 29425 Office: 843-792-2153 Lab: 843-792-3530 Fax: 843-792-8436 Email: [hidden email] http://academicdepartments.musc.edu/ccdir -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Kate Luby-Phelps Sent: Saturday, April 23, 2011 11:22 AM To: [hidden email] Subject: Re: Live cell imaging with multiphoton confocal ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Multiphoton has no clear advantage for cells in culture. For cell culture samples, we use two photon only to excite DAPI or other UV and near-UV fluors since we had to make a choice between the Ti:S and a 405 laser on our scope. Kate Luby-Phelps |
|
Mark Cannell |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Err, is a Ti:S not a very "specialized laser" ? ;-) > > Multiphoton microscopy is helpful to excite UV-absorbing > fluorophores like NAD(P)H that cannot otherwise be excited without > specialized lasers and optics. > |
|
Knecht, David |
|
|
In reply to this post by Kate Luby-Phelps
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Is there any disadvantage to multiphoton for cultured cells (besides cost and complexity)? Cell viability? Phototoxicity? Dave On Apr 23, 2011, at 11:21 AM, Kate Luby-Phelps wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Multiphoton has no clear advantage for cells in culture. For cell culture samples, > we use two photon only to excite DAPI or other UV and near-UV fluors since we > had to make a choice between the Ti:S and a 405 laser on our scope. > > Kate Luby-Phelps Dr. David Knecht Department of Molecular and Cell Biology Co-head Flow Cytometry and Confocal Microscopy Facility U-3125 91 N. Eagleville Rd. University of Connecticut Storrs, CT 06269 860-486-2200 860-486-4331 (fax) |
|
Lemasters, John J. |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi David and Mark, Good points. By specialized laser and optics, I meant a UV argon laser and UV quartz optics. I had such a beast once to image NADH, but multiphoton is much easier and more versatile. Regarding multiphoton photodamage, David Piston has a paper showing that 2 photon imaging is more damaging than 1 photon imaging for cells in culture and that the multiphoton photodamage has a 3 photon power profile. Hence most folks will tell you to use conventional 1 photon imaging for cells in culture -- but you'll still need 2-photon microscopy to image things like NADH and NADPH unless, of course, you have a UV laser and UV optics. John -- John J. Lemasters, MD, PhD Professor and GlaxoSmithKline Distinguished Endowed Chair Director, Center for Cell Death, Injury and Regeneration Departments of Pharmaceutical & Biomedical Sciences and Biochemistry & Molecular Biology Medical University of South Carolina QF213 Quadrangle Building 280 Calhoun Street, MSC 140 Charleston, SC 29425 Office: 843-792-2153 Lab: 843-792-3530 Fax: 843-792-8436 Email: [hidden email] http://academicdepartments.musc.edu/ccdir -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of David Knecht charter Sent: Saturday, April 23, 2011 6:03 PM To: [hidden email] Subject: Re: Live cell imaging with multiphoton confocal ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Is there any disadvantage to multiphoton for cultured cells (besides cost and complexity)? Cell viability? Phototoxicity? Dave On Apr 23, 2011, at 11:21 AM, Kate Luby-Phelps wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Multiphoton has no clear advantage for cells in culture. For cell > culture samples, we use two photon only to excite DAPI or other UV > and near-UV fluors since we had to make a choice between the Ti:S and a 405 laser on our scope. > > Kate Luby-Phelps Dr. David Knecht Department of Molecular and Cell Biology Co-head Flow Cytometry and Confocal Microscopy Facility U-3125 91 N. Eagleville Rd. University of Connecticut Storrs, CT 06269 860-486-2200 860-486-4331 (fax) |
|
Lemasters, John J. |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** P.S. UV-transmitting optics don't have to be quartz. Many lens used for UV illumination of Fura2 or Indo1 transmit UV but are still not chromatically corrected in the UV as is desirable if not necessary for point scanning confocal. -- John J. Lemasters, MD, PhD Professor and GlaxoSmithKline Distinguished Endowed Chair Director, Center for Cell Death, Injury and Regeneration Departments of Pharmaceutical & Biomedical Sciences and Biochemistry & Molecular Biology Medical University of South Carolina QF213 Quadrangle Building 280 Calhoun Street, MSC 140 Charleston, SC 29425 Office: 843-792-2153 Lab: 843-792-3530 Fax: 843-792-8436 Email: [hidden email] http://academicdepartments.musc.edu/ccdir -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Lemasters, John Sent: Saturday, April 23, 2011 9:21 PM To: [hidden email] Subject: Re: Live cell imaging with multiphoton confocal ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi David and Mark, Good points. By specialized laser and optics, I meant a UV argon laser and UV quartz optics. I had such a beast once to image NADH, but multiphoton is much easier and more versatile. Regarding multiphoton photodamage, David Piston has a paper showing that 2 photon imaging is more damaging than 1 photon imaging for cells in culture and that the multiphoton photodamage has a 3 photon power profile. Hence most folks will tell you to use conventional 1 photon imaging for cells in culture -- but you'll still need 2-photon microscopy to image things like NADH and NADPH unless, of course, you have a UV laser and UV optics. John -- John J. Lemasters, MD, PhD Professor and GlaxoSmithKline Distinguished Endowed Chair Director, Center for Cell Death, Injury and Regeneration Departments of Pharmaceutical & Biomedical Sciences and Biochemistry & Molecular Biology Medical University of South Carolina QF213 Quadrangle Building 280 Calhoun Street, MSC 140 Charleston, SC 29425 Office: 843-792-2153 Lab: 843-792-3530 Fax: 843-792-8436 Email: [hidden email] http://academicdepartments.musc.edu/ccdir -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of David Knecht charter Sent: Saturday, April 23, 2011 6:03 PM To: [hidden email] Subject: Re: Live cell imaging with multiphoton confocal ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Is there any disadvantage to multiphoton for cultured cells (besides cost and complexity)? Cell viability? Phototoxicity? Dave On Apr 23, 2011, at 11:21 AM, Kate Luby-Phelps wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Multiphoton has no clear advantage for cells in culture. For cell > culture samples, we use two photon only to excite DAPI or other UV > and near-UV fluors since we had to make a choice between the Ti:S and a 405 laser on our scope. > > Kate Luby-Phelps Dr. David Knecht Department of Molecular and Cell Biology Co-head Flow Cytometry and Confocal Microscopy Facility U-3125 91 N. Eagleville Rd. University of Connecticut Storrs, CT 06269 860-486-2200 860-486-4331 (fax) |
|
George McNamara |
|
|
In reply to this post by Knecht, David
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi David, The tuning range of Ti:Sapphire (ex. 690-1040 nm) used to be problematic for exciting far red and infrared fluorophores. There were some exceptions, such as Alexa Fluor 488m 568, 594, and 633 all exciting well around 755 nm (Fig 5D of Dickinson et al 2003 J Biomed Optics 8: 329-338), but since you are ok with cost and complexity, adding an OPO(s) and/or multiple MP lasers, gives you full range (see http://www.coherent.com/products/?1570/Chameleon-OPO-Tuning-Range for OPO). Depending on cell type and physiological needs, you may be able to suppress phototoxicity by lowering O2 - cell culture at ~18% O2 is something of an artifact for many cell types - by using catalase/glucose oxidase/glucose or Oxyrase (for the latter, www.oxyrase.com, see Waterman-Storer 1993 and Mikhailov 1995 Cell Motil Cytokseleton). Enjoy, George On 4/23/2011 6:03 PM, David Knecht charter wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Is there any disadvantage to multiphoton for cultured cells (besides cost and complexity)? Cell viability? Phototoxicity? Dave > > On Apr 23, 2011, at 11:21 AM, Kate Luby-Phelps wrote: > > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Multiphoton has no clear advantage for cells in culture. For cell culture samples, >> we use two photon only to excite DAPI or other UV and near-UV fluors since we >> had to make a choice between the Ti:S and a 405 laser on our scope. >> >> Kate Luby-Phelps >> > Dr. David Knecht > Department of Molecular and Cell Biology > Co-head Flow Cytometry and Confocal Microscopy Facility > U-3125 > 91 N. Eagleville Rd. > University of Connecticut > Storrs, CT 06269 > 860-486-2200 > 860-486-4331 (fax) > > -- George McNamara, PhD Analytical Imaging Core Facility University of Miami |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello - Question for those of you who have a Leica SP5X (the white light laser version) - What has been your experience with maintenance and need for service on the white light laser? Do people recommend getting the service contract on the laser once the warrantee at purchase is over? I run a microscopy facility and have service contracts on all the equipment and do feel it's a good idea especially in a facility. However Leica has made the white light laser service contract separate from the main confocal service contract and the price of both more than doubles the yearly cost of the SP5 alone. Thanks for any input (and feel free to contact me off list). - Lisa --------------------------------------- Lisa Cameron Dana Farber Cancer Institute Boston, MA [hidden email] The information in this e-mail is intended only for the person to whom it is addressed. If you believe this e-mail was sent to you in error and the e-mail contains patient information, please contact the Partners Compliance HelpLine at http://www.partners.org/complianceline . If the e-mail was sent to you in error but does not contain patient information, please contact the sender and properly dispose of the e-mail. |
|
Craig Brideau |
|
|
In reply to this post by George McNamara
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I second the previous lister's opinions. At the end of the day MP is really only good for thick tissue sections. If you are looking at cell layers just use a conventional 1P confocal. The key advantage of 2P is penetration depth, which is moot when you are looking at cells. That said, John's comments about shorter UV imaging are still valid; it's easier to get a NIR Ti:Saph beam through conventional optics than a very UV beam. This only applies if you are using dyes that need very UV excitation of course. Craig On Sun, Apr 24, 2011 at 6:29 AM, George McNamara <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi David, > > The tuning range of Ti:Sapphire (ex. 690-1040 nm) used to be problematic > for exciting far red and infrared fluorophores. There were some exceptions, > such as Alexa Fluor 488m 568, 594, and 633 all exciting well around 755 nm > (Fig 5D of Dickinson et al 2003 J Biomed Optics 8: 329-338), but since you > are ok with cost and complexity, adding an OPO(s) and/or multiple MP lasers, > gives you full range (see > http://www.coherent.com/products/?1570/Chameleon-OPO-Tuning-Range for > OPO). > > Depending on cell type and physiological needs, you may be able to suppress > phototoxicity by lowering O2 - cell culture at ~18% O2 is something of an > artifact for many cell types - by using catalase/glucose oxidase/glucose or > Oxyrase (for the latter, www.oxyrase.com, see Waterman-Storer 1993 and > Mikhailov 1995 Cell Motil Cytokseleton). > > Enjoy, > > George > > > On 4/23/2011 6:03 PM, David Knecht charter wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Is there any disadvantage to multiphoton for cultured cells (besides cost >> and complexity)? Cell viability? Phototoxicity? Dave >> >> On Apr 23, 2011, at 11:21 AM, Kate Luby-Phelps wrote: >> >> >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Multiphoton has no clear advantage for cells in culture. For cell culture >>> samples, >>> we use two photon only to excite DAPI or other UV and near-UV fluors >>> since we >>> had to make a choice between the Ti:S and a 405 laser on our scope. >>> >>> Kate Luby-Phelps >>> >>> >> Dr. David Knecht >> Department of Molecular and Cell Biology >> Co-head Flow Cytometry and Confocal Microscopy Facility >> U-3125 >> 91 N. Eagleville Rd. >> University of Connecticut >> Storrs, CT 06269 >> 860-486-2200 >> 860-486-4331 (fax) >> >> >> > > > -- > > > George McNamara, PhD > Analytical Imaging Core Facility > University of Miami > |
|
Rosemary.White |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear all, Haven't followed this discussion closely, but would one advantage of 2P lie in the ability to uncage caged compounds - fluorescent tracers, for example, within a single cell? We've used UV uncaging and while you get most uncaging in the target cell, you also get cones of uncaging above and below the plane of focus. Seems that 2P would overcome this problem. Rosemary White Dr Rosemary White CSIRO Plant Industry GPO Box 1600 Canberra, ACT 2601 Australia T 61 2 6246 5475 F 61 2 6246 5334 E [hidden email] On 28/04/11 10:10 AM, "Craig Brideau" <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I second the previous lister's opinions. At the end of the day MP is really > only good for thick tissue sections. If you are looking at cell layers just > use a conventional 1P confocal. The key advantage of 2P is penetration > depth, which is moot when you are looking at cells. That said, John's > comments about shorter UV imaging are still valid; it's easier to get a NIR > Ti:Saph beam through conventional optics than a very UV beam. This only > applies if you are using dyes that need very UV excitation of course. > > Craig > > > > > On Sun, Apr 24, 2011 at 6:29 AM, George McNamara > <[hidden email]>wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hi David, >> >> The tuning range of Ti:Sapphire (ex. 690-1040 nm) used to be problematic >> for exciting far red and infrared fluorophores. There were some exceptions, >> such as Alexa Fluor 488m 568, 594, and 633 all exciting well around 755 nm >> (Fig 5D of Dickinson et al 2003 J Biomed Optics 8: 329-338), but since you >> are ok with cost and complexity, adding an OPO(s) and/or multiple MP lasers, >> gives you full range (see >> http://www.coherent.com/products/?1570/Chameleon-OPO-Tuning-Range for >> OPO). >> >> Depending on cell type and physiological needs, you may be able to suppress >> phototoxicity by lowering O2 - cell culture at ~18% O2 is something of an >> artifact for many cell types - by using catalase/glucose oxidase/glucose or >> Oxyrase (for the latter, www.oxyrase.com, see Waterman-Storer 1993 and >> Mikhailov 1995 Cell Motil Cytokseleton). >> >> Enjoy, >> >> George >> >> >> On 4/23/2011 6:03 PM, David Knecht charter wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Is there any disadvantage to multiphoton for cultured cells (besides cost >>> and complexity)? Cell viability? Phototoxicity? Dave >>> >>> On Apr 23, 2011, at 11:21 AM, Kate Luby-Phelps wrote: >>> >>> >>> >>>> ***** >>>> To join, leave or search the confocal microscopy listserv, go to: >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>> ***** >>>> >>>> Multiphoton has no clear advantage for cells in culture. For cell culture >>>> samples, >>>> we use two photon only to excite DAPI or other UV and near-UV fluors >>>> since we >>>> had to make a choice between the Ti:S and a 405 laser on our scope. >>>> >>>> Kate Luby-Phelps >>>> >>>> >>> Dr. David Knecht >>> Department of Molecular and Cell Biology >>> Co-head Flow Cytometry and Confocal Microscopy Facility >>> U-3125 >>> 91 N. Eagleville Rd. >>> University of Connecticut >>> Storrs, CT 06269 >>> 860-486-2200 >>> 860-486-4331 (fax) >>> >>> >>> >> >> >> -- >> >> >> George McNamara, PhD >> Analytical Imaging Core Facility >> University of Miami >> |
|
Mark Cannell |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Actually, that is a very good use of 2P and it works really well (shameless self citation follows) See: Cannell, M.B., Jacobs, M., Donaldson, P. & Soeller, C. (2004) Probing microscopic diffusion by 2-photon flash photolysis: Visualization of isotropic and anisotropic diffusion between lens fiber cells. Micr. Res. Tech. 63(1) 50-57 Soeller, C., Jacobs, M.D., Jones, K.T., Ellis-Davies, G.C.R., Donaldson, P.J. and Cannell, M.B. (2003) Probing microscopic transport and release fluxes in cells by two-photon flash photolysis. J. Biomed. Opt. 8:418-427 Soeller, C. & Cannell, M.B. (2002) Estimation of the sarcoplasmic reticulum Ca2+ release flux underlying Ca2+ sparks. Biophys. J. 82. 2396-2414. Cheers On 28/04/2011, at 12:48 PM, Rosemary White wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all, > > Haven't followed this discussion closely, but would one advantage of > 2P lie > in the ability to uncage caged compounds - fluorescent tracers, for > example, > within a single cell? We've used UV uncaging and while you get most > uncaging in the target cell, you also get cones of uncaging above > and below > the plane of focus. Seems that 2P would overcome this problem. > > Rosemary White > > Dr Rosemary White > CSIRO Plant Industry > GPO Box 1600 > Canberra, ACT 2601 > Australia > > T 61 2 6246 5475 > F 61 2 6246 5334 > E [hidden email] > > > > On 28/04/11 10:10 AM, "Craig Brideau" <[hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> I second the previous lister's opinions. At the end of the day MP >> is really >> only good for thick tissue sections. If you are looking at cell >> layers just >> use a conventional 1P confocal. The key advantage of 2P is >> penetration >> depth, which is moot when you are looking at cells. That said, >> John's >> comments about shorter UV imaging are still valid; it's easier to >> get a NIR >> Ti:Saph beam through conventional optics than a very UV beam. This >> only >> applies if you are using dyes that need very UV excitation of course. >> >> Craig >> >> >> >> >> On Sun, Apr 24, 2011 at 6:29 AM, George McNamara >> <[hidden email]>wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Hi David, >>> >>> The tuning range of Ti:Sapphire (ex. 690-1040 nm) used to be >>> problematic >>> for exciting far red and infrared fluorophores. There were some >>> exceptions, >>> such as Alexa Fluor 488m 568, 594, and 633 all exciting well >>> around 755 nm >>> (Fig 5D of Dickinson et al 2003 J Biomed Optics 8: 329-338), but >>> since you >>> are ok with cost and complexity, adding an OPO(s) and/or multiple >>> MP lasers, >>> gives you full range (see >>> http://www.coherent.com/products/?1570/Chameleon-OPO-Tuning-Range >>> for >>> OPO). >>> >>> Depending on cell type and physiological needs, you may be able to >>> suppress >>> phototoxicity by lowering O2 - cell culture at ~18% O2 is >>> something of an >>> artifact for many cell types - by using catalase/glucose oxidase/ >>> glucose or >>> Oxyrase (for the latter, www.oxyrase.com, see Waterman-Storer 1993 >>> and >>> Mikhailov 1995 Cell Motil Cytokseleton). >>> >>> Enjoy, >>> >>> George >>> >>> >>> On 4/23/2011 6:03 PM, David Knecht charter wrote: >>> >>>> ***** >>>> To join, leave or search the confocal microscopy listserv, go to: >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>> ***** >>>> >>>> Is there any disadvantage to multiphoton for cultured cells >>>> (besides cost >>>> and complexity)? Cell viability? Phototoxicity? Dave >>>> >>>> On Apr 23, 2011, at 11:21 AM, Kate Luby-Phelps wrote: >>>> >>>> >>>> >>>>> ***** >>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>> ***** >>>>> >>>>> Multiphoton has no clear advantage for cells in culture. For >>>>> cell culture >>>>> samples, >>>>> we use two photon only to excite DAPI or other UV and near-UV >>>>> fluors >>>>> since we >>>>> had to make a choice between the Ti:S and a 405 laser on our >>>>> scope. >>>>> >>>>> Kate Luby-Phelps >>>>> >>>>> >>>> Dr. David Knecht >>>> Department of Molecular and Cell Biology >>>> Co-head Flow Cytometry and Confocal Microscopy Facility >>>> U-3125 >>>> 91 N. Eagleville Rd. >>>> University of Connecticut >>>> Storrs, CT 06269 >>>> 860-486-2200 >>>> 860-486-4331 (fax) >>>> >>>> >>>> >>> >>> >>> -- >>> >>> >>> George McNamara, PhD >>> Analytical Imaging Core Facility >>> University of Miami >>> |
|
Craig Brideau |
|
|
In reply to this post by Rosemary.White
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** How much uncaging occurs outside the focal region in single photon? If you don't have your laser turned up to much the energy density should only exceed the uncaging threshold near the focal point, yes? Craig On Wed, Apr 27, 2011 at 6:48 PM, Rosemary White <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all, > > Haven't followed this discussion closely, but would one advantage of 2P lie > in the ability to uncage caged compounds - fluorescent tracers, for > example, > within a single cell? We've used UV uncaging and while you get most > uncaging in the target cell, you also get cones of uncaging above and below > the plane of focus. Seems that 2P would overcome this problem. > > Rosemary White > > Dr Rosemary White > CSIRO Plant Industry > GPO Box 1600 > Canberra, ACT 2601 > Australia > > T 61 2 6246 5475 > F 61 2 6246 5334 > E [hidden email] > > > > On 28/04/11 10:10 AM, "Craig Brideau" <[hidden email]> wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > I second the previous lister's opinions. At the end of the day MP is > really > > only good for thick tissue sections. If you are looking at cell layers > just > > use a conventional 1P confocal. The key advantage of 2P is penetration > > depth, which is moot when you are looking at cells. That said, John's > > comments about shorter UV imaging are still valid; it's easier to get a > NIR > > Ti:Saph beam through conventional optics than a very UV beam. This only > > applies if you are using dyes that need very UV excitation of course. > > > > Craig > > > > > > > > > > On Sun, Apr 24, 2011 at 6:29 AM, George McNamara > > <[hidden email]>wrote: > > > >> ***** > >> To join, leave or search the confocal microscopy listserv, go to: > >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >> ***** > >> > >> Hi David, > >> > >> The tuning range of Ti:Sapphire (ex. 690-1040 nm) used to be problematic > >> for exciting far red and infrared fluorophores. There were some > exceptions, > >> such as Alexa Fluor 488m 568, 594, and 633 all exciting well around 755 > nm > >> (Fig 5D of Dickinson et al 2003 J Biomed Optics 8: 329-338), but since > you > >> are ok with cost and complexity, adding an OPO(s) and/or multiple MP > lasers, > >> gives you full range (see > >> http://www.coherent.com/products/?1570/Chameleon-OPO-Tuning-Range for > >> OPO). > >> > >> Depending on cell type and physiological needs, you may be able to > suppress > >> phototoxicity by lowering O2 - cell culture at ~18% O2 is something of > an > >> artifact for many cell types - by using catalase/glucose oxidase/glucose > or > >> Oxyrase (for the latter, www.oxyrase.com, see Waterman-Storer 1993 and > >> Mikhailov 1995 Cell Motil Cytokseleton). > >> > >> Enjoy, > >> > >> George > >> > >> > >> On 4/23/2011 6:03 PM, David Knecht charter wrote: > >> > >>> ***** > >>> To join, leave or search the confocal microscopy listserv, go to: > >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >>> ***** > >>> > >>> Is there any disadvantage to multiphoton for cultured cells (besides > cost > >>> and complexity)? Cell viability? Phototoxicity? Dave > >>> > >>> On Apr 23, 2011, at 11:21 AM, Kate Luby-Phelps wrote: > >>> > >>> > >>> > >>>> ***** > >>>> To join, leave or search the confocal microscopy listserv, go to: > >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >>>> ***** > >>>> > >>>> Multiphoton has no clear advantage for cells in culture. For cell > culture > >>>> samples, > >>>> we use two photon only to excite DAPI or other UV and near-UV fluors > >>>> since we > >>>> had to make a choice between the Ti:S and a 405 laser on our scope. > >>>> > >>>> Kate Luby-Phelps > >>>> > >>>> > >>> Dr. David Knecht > >>> Department of Molecular and Cell Biology > >>> Co-head Flow Cytometry and Confocal Microscopy Facility > >>> U-3125 > >>> 91 N. Eagleville Rd. > >>> University of Connecticut > >>> Storrs, CT 06269 > >>> 860-486-2200 > >>> 860-486-4331 (fax) > >>> > >>> > >>> > >> > >> > >> -- > >> > >> > >> George McNamara, PhD > >> Analytical Imaging Core Facility > >> University of Miami > >> > |
|
Mark Cannell |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** At low powers the uncaging is linear in intensity so there should be no threshold. So, the uncaging volume at low powers is the same as the PSF and for 2P its the 2P PSF. If ground state depletion develops the FWHM of the focal uncaging spot grows (of course). Mark On 28/04/2011, at 2:18 PM, Craig Brideau wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > How much uncaging occurs outside the focal region in single photon? > If you > don't have your laser turned up to much the energy density should only > exceed the uncaging threshold near the focal point, yes? > > Craig > > > On Wed, Apr 27, 2011 at 6:48 PM, Rosemary White <[hidden email] > >wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear all, >> >> Haven't followed this discussion closely, but would one advantage >> of 2P lie >> in the ability to uncage caged compounds - fluorescent tracers, for >> example, >> within a single cell? We've used UV uncaging and while you get most >> uncaging in the target cell, you also get cones of uncaging above >> and below >> the plane of focus. Seems that 2P would overcome this problem. >> >> Rosemary White >> >> Dr Rosemary White >> CSIRO Plant Industry >> GPO Box 1600 >> Canberra, ACT 2601 >> Australia >> >> T 61 2 6246 5475 >> F 61 2 6246 5334 >> E [hidden email] >> >> >> >> On 28/04/11 10:10 AM, "Craig Brideau" <[hidden email]> >> wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> I second the previous lister's opinions. At the end of the day MP >>> is >> really >>> only good for thick tissue sections. If you are looking at cell >>> layers >> just >>> use a conventional 1P confocal. The key advantage of 2P is >>> penetration >>> depth, which is moot when you are looking at cells. That said, >>> John's >>> comments about shorter UV imaging are still valid; it's easier to >>> get a >> NIR >>> Ti:Saph beam through conventional optics than a very UV beam. >>> This only >>> applies if you are using dyes that need very UV excitation of >>> course. >>> >>> Craig >>> >>> >>> >>> >>> On Sun, Apr 24, 2011 at 6:29 AM, George McNamara >>> <[hidden email]>wrote: >>> >>>> ***** >>>> To join, leave or search the confocal microscopy listserv, go to: >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>> ***** >>>> >>>> Hi David, >>>> >>>> The tuning range of Ti:Sapphire (ex. 690-1040 nm) used to be >>>> problematic >>>> for exciting far red and infrared fluorophores. There were some >> exceptions, >>>> such as Alexa Fluor 488m 568, 594, and 633 all exciting well >>>> around 755 >> nm >>>> (Fig 5D of Dickinson et al 2003 J Biomed Optics 8: 329-338), but >>>> since >> you >>>> are ok with cost and complexity, adding an OPO(s) and/or multiple >>>> MP >> lasers, >>>> gives you full range (see >>>> http://www.coherent.com/products/?1570/Chameleon-OPO-Tuning-Range >>>> for >>>> OPO). >>>> >>>> Depending on cell type and physiological needs, you may be able to >> suppress >>>> phototoxicity by lowering O2 - cell culture at ~18% O2 is >>>> something of >> an >>>> artifact for many cell types - by using catalase/glucose oxidase/ >>>> glucose >> or >>>> Oxyrase (for the latter, www.oxyrase.com, see Waterman-Storer >>>> 1993 and >>>> Mikhailov 1995 Cell Motil Cytokseleton). >>>> >>>> Enjoy, >>>> >>>> George >>>> >>>> >>>> On 4/23/2011 6:03 PM, David Knecht charter wrote: >>>> >>>>> ***** >>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>> ***** >>>>> >>>>> Is there any disadvantage to multiphoton for cultured cells >>>>> (besides >> cost >>>>> and complexity)? Cell viability? Phototoxicity? Dave >>>>> >>>>> On Apr 23, 2011, at 11:21 AM, Kate Luby-Phelps wrote: >>>>> >>>>> >>>>> >>>>>> ***** >>>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>>> ***** >>>>>> >>>>>> Multiphoton has no clear advantage for cells in culture. For cell >> culture >>>>>> samples, >>>>>> we use two photon only to excite DAPI or other UV and near-UV >>>>>> fluors >>>>>> since we >>>>>> had to make a choice between the Ti:S and a 405 laser on our >>>>>> scope. >>>>>> >>>>>> Kate Luby-Phelps >>>>>> >>>>>> >>>>> Dr. David Knecht >>>>> Department of Molecular and Cell Biology >>>>> Co-head Flow Cytometry and Confocal Microscopy Facility >>>>> U-3125 >>>>> 91 N. Eagleville Rd. >>>>> University of Connecticut >>>>> Storrs, CT 06269 >>>>> 860-486-2200 >>>>> 860-486-4331 (fax) >>>>> >>>>> >>>>> >>>> >>>> >>>> -- >>>> >>>> >>>> George McNamara, PhD >>>> Analytical Imaging Core Facility >>>> University of Miami >>>> >> |
|
Sudipta Maiti |
|
|
In reply to this post by Craig Brideau
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Other than UV dyes, native chromophores , such as NaDh, serotoni etc., are only accessible through multiphoton for live cell UV imaging. I can send a large list of our work in this area (off the list)if anyone cares to know about it. Sudipta which are On Wed, 27 Apr 2011 18:10:16 -0600, Craig Brideau wrote > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I second the previous lister's opinions. At the end of the day MP > is really only good for thick tissue sections. If you are looking > at cell layers just use a conventional 1P confocal. The key > advantage of 2P is penetration depth, which is moot when you are > looking at cells. That said, John's comments about shorter UV > imaging are still valid; it's easier to get a NIR Ti:Saph beam > through conventional optics than a very UV beam. This only applies > if you are using dyes that need very UV excitation of course. > > Craig > > On Sun, Apr 24, 2011 at 6:29 AM, George McNamara > <[hidden email]>wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Hi David, > > > > The tuning range of Ti:Sapphire (ex. 690-1040 nm) used to be problematic > > for exciting far red and infrared fluorophores. There were some > > such as Alexa Fluor 488m 568, 594, and 633 all exciting well around 755 nm > > (Fig 5D of Dickinson et al 2003 J Biomed Optics 8: 329-338), but since you > > are ok with cost and complexity, adding an OPO(s) and/or multiple MP lasers, > > gives you full range (see > > http://www.coherent.com/products/?1570/Chameleon-OPO-Tuning-Range for > > OPO). > > > > Depending on cell type and physiological needs, you may be able to suppress > > phototoxicity by lowering O2 - cell culture at ~18% O2 is something of an > > artifact for many cell types - by using catalase/glucose oxidase/glucose or > > Oxyrase (for the latter, www.oxyrase.com, see Waterman-Storer 1993 and > > Mikhailov 1995 Cell Motil Cytokseleton). > > > > Enjoy, > > > > George > > > > > > On 4/23/2011 6:03 PM, David Knecht charter wrote: > > > >> ***** > >> To join, leave or search the confocal microscopy listserv, go to: > >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >> ***** > >> > >> Is there any disadvantage to multiphoton for cultured cells (besides cost > >> and complexity)? Cell viability? Phototoxicity? Dave > >> > >> On Apr 23, 2011, at 11:21 AM, Kate Luby-Phelps wrote: > >> > >> > >> > >>> ***** > >>> To join, leave or search the confocal microscopy listserv, go to: > >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >>> ***** > >>> > >>> Multiphoton has no clear advantage for cells in culture. For cell > >>> samples, > >>> we use two photon only to excite DAPI or other UV and near-UV fluors > >>> since we > >>> had to make a choice between the Ti:S and a 405 laser on our scope. > >>> > >>> Kate Luby-Phelps > >>> > >>> > >> Dr. David Knecht > >> Department of Molecular and Cell Biology > >> Co-head Flow Cytometry and Confocal Microscopy Facility > >> U-3125 > >> 91 N. Eagleville Rd. > >> University of Connecticut > >> Storrs, CT 06269 > >> 860-486-2200 > >> 860-486-4331 (fax) > >> > >> > >> > > > > > > -- > > > > > > George McNamara, PhD > > Analytical Imaging Core Facility > > University of Miami > > Dr. Sudipta Maiti Associate Professor Dept. of Chemical Sciences Tata Institute of Fundamental Research Homi Bhabha Raod, Colaba, Mumbai 400005 Ph. 91-22-2278-2716 / 2539 Fax: 91-22-2280-4610 alternate e-mail: [hidden email] url: biophotonics.wetpaint.com |
|
Craig Brideau |
|
|
In reply to this post by Mark Cannell
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** So if it is linear with intensity (1P), should it roll off to 1/4 in terms of distance from the ideal focal point? Intensity drops to 1/4 at 1 unit distance from a point source, so I'm approximating the center of the focal point here. Am I right in my thinking? Craig On Wed, Apr 27, 2011 at 8:38 PM, Mark Cannell <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > At low powers the uncaging is linear in intensity so there should be no > threshold. So, the uncaging volume at low powers is the same as the PSF and > for 2P its the 2P PSF. If ground state depletion develops the FWHM of the > focal uncaging spot grows (of course). > > Mark > > > On 28/04/2011, at 2:18 PM, Craig Brideau wrote: > > ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> How much uncaging occurs outside the focal region in single photon? If >> you >> don't have your laser turned up to much the energy density should only >> exceed the uncaging threshold near the focal point, yes? >> >> Craig >> >> >> On Wed, Apr 27, 2011 at 6:48 PM, Rosemary White <[hidden email] >> >wrote: >> >> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Dear all, >>> >>> Haven't followed this discussion closely, but would one advantage of 2P >>> lie >>> in the ability to uncage caged compounds - fluorescent tracers, for >>> example, >>> within a single cell? We've used UV uncaging and while you get most >>> uncaging in the target cell, you also get cones of uncaging above and >>> below >>> the plane of focus. Seems that 2P would overcome this problem. >>> >>> Rosemary White >>> >>> Dr Rosemary White >>> CSIRO Plant Industry >>> GPO Box 1600 >>> Canberra, ACT 2601 >>> Australia >>> >>> T 61 2 6246 5475 >>> F 61 2 6246 5334 >>> E [hidden email] >>> >>> >>> >>> On 28/04/11 10:10 AM, "Craig Brideau" <[hidden email]> wrote: >>> >>> ***** >>>> To join, leave or search the confocal microscopy listserv, go to: >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>> ***** >>>> >>>> I second the previous lister's opinions. At the end of the day MP is >>>> >>> really >>> >>>> only good for thick tissue sections. If you are looking at cell layers >>>> >>> just >>> >>>> use a conventional 1P confocal. The key advantage of 2P is penetration >>>> depth, which is moot when you are looking at cells. That said, John's >>>> comments about shorter UV imaging are still valid; it's easier to get a >>>> >>> NIR >>> >>>> Ti:Saph beam through conventional optics than a very UV beam. This only >>>> applies if you are using dyes that need very UV excitation of course. >>>> >>>> Craig >>>> >>>> >>>> >>>> >>>> On Sun, Apr 24, 2011 at 6:29 AM, George McNamara >>>> <[hidden email]>wrote: >>>> >>>> ***** >>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>> ***** >>>>> >>>>> Hi David, >>>>> >>>>> The tuning range of Ti:Sapphire (ex. 690-1040 nm) used to be >>>>> problematic >>>>> for exciting far red and infrared fluorophores. There were some >>>>> >>>> exceptions, >>> >>>> such as Alexa Fluor 488m 568, 594, and 633 all exciting well around 755 >>>>> >>>> nm >>> >>>> (Fig 5D of Dickinson et al 2003 J Biomed Optics 8: 329-338), but since >>>>> >>>> you >>> >>>> are ok with cost and complexity, adding an OPO(s) and/or multiple MP >>>>> >>>> lasers, >>> >>>> gives you full range (see >>>>> http://www.coherent.com/products/?1570/Chameleon-OPO-Tuning-Range for >>>>> OPO). >>>>> >>>>> Depending on cell type and physiological needs, you may be able to >>>>> >>>> suppress >>> >>>> phototoxicity by lowering O2 - cell culture at ~18% O2 is something of >>>>> >>>> an >>> >>>> artifact for many cell types - by using catalase/glucose oxidase/glucose >>>>> >>>> or >>> >>>> Oxyrase (for the latter, www.oxyrase.com, see Waterman-Storer 1993 and >>>>> Mikhailov 1995 Cell Motil Cytokseleton). >>>>> >>>>> Enjoy, >>>>> >>>>> George >>>>> >>>>> >>>>> On 4/23/2011 6:03 PM, David Knecht charter wrote: >>>>> >>>>> ***** >>>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>>> ***** >>>>>> >>>>>> Is there any disadvantage to multiphoton for cultured cells (besides >>>>>> >>>>> cost >>> >>>> and complexity)? Cell viability? Phototoxicity? Dave >>>>>> >>>>>> On Apr 23, 2011, at 11:21 AM, Kate Luby-Phelps wrote: >>>>>> >>>>>> >>>>>> >>>>>> ***** >>>>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>>>> ***** >>>>>>> >>>>>>> Multiphoton has no clear advantage for cells in culture. For cell >>>>>>> >>>>>> culture >>> >>>> samples, >>>>>>> we use two photon only to excite DAPI or other UV and near-UV fluors >>>>>>> since we >>>>>>> had to make a choice between the Ti:S and a 405 laser on our scope. >>>>>>> >>>>>>> Kate Luby-Phelps >>>>>>> >>>>>>> >>>>>>> Dr. David Knecht >>>>>> Department of Molecular and Cell Biology >>>>>> Co-head Flow Cytometry and Confocal Microscopy Facility >>>>>> U-3125 >>>>>> 91 N. Eagleville Rd. >>>>>> University of Connecticut >>>>>> Storrs, CT 06269 >>>>>> 860-486-2200 >>>>>> 860-486-4331 (fax) >>>>>> >>>>>> >>>>>> >>>>>> >>>>> >>>>> -- >>>>> >>>>> >>>>> George McNamara, PhD >>>>> Analytical Imaging Core Facility >>>>> University of Miami >>>>> >>>>> >>> |
|
Mark Cannell |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Yes but the area is increased by a similar amount so the total released does not fall with distance (assuming no inner filtering effects etc) Cheers On 29/04/2011, at 7:41 AM, Craig Brideau wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > So if it is linear with intensity (1P), should it roll off to 1/4 in > terms > of distance from the ideal focal point? Intensity drops to 1/4 at 1 > unit > distance from a point source, so I'm approximating the center of the > focal > point here. Am I right in my thinking? > > Craig > > > > On Wed, Apr 27, 2011 at 8:38 PM, Mark Cannell <[hidden email] > >wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> At low powers the uncaging is linear in intensity so there should >> be no >> threshold. So, the uncaging volume at low powers is the same as >> the PSF and >> for 2P its the 2P PSF. If ground state depletion develops the FWHM >> of the >> focal uncaging spot grows (of course). >> >> Mark >> >> >> On 28/04/2011, at 2:18 PM, Craig Brideau wrote: >> >> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> How much uncaging occurs outside the focal region in single >>> photon? If >>> you >>> don't have your laser turned up to much the energy density should >>> only >>> exceed the uncaging threshold near the focal point, yes? >>> >>> Craig >>> >>> >>> On Wed, Apr 27, 2011 at 6:48 PM, Rosemary White <[hidden email] >>>> wrote: >>> >>> ***** >>>> To join, leave or search the confocal microscopy listserv, go to: >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>> ***** >>>> >>>> Dear all, >>>> >>>> Haven't followed this discussion closely, but would one advantage >>>> of 2P >>>> lie >>>> in the ability to uncage caged compounds - fluorescent tracers, for >>>> example, >>>> within a single cell? We've used UV uncaging and while you get >>>> most >>>> uncaging in the target cell, you also get cones of uncaging above >>>> and >>>> below >>>> the plane of focus. Seems that 2P would overcome this problem. >>>> >>>> Rosemary White >>>> >>>> Dr Rosemary White >>>> CSIRO Plant Industry >>>> GPO Box 1600 >>>> Canberra, ACT 2601 >>>> Australia >>>> >>>> T 61 2 6246 5475 >>>> F 61 2 6246 5334 >>>> E [hidden email] >>>> >>>> >>>> >>>> On 28/04/11 10:10 AM, "Craig Brideau" <[hidden email]> >>>> wrote: >>>> >>>> ***** >>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>> ***** >>>>> >>>>> I second the previous lister's opinions. At the end of the day >>>>> MP is >>>>> >>>> really >>>> >>>>> only good for thick tissue sections. If you are looking at cell >>>>> layers >>>>> >>>> just >>>> >>>>> use a conventional 1P confocal. The key advantage of 2P is >>>>> penetration >>>>> depth, which is moot when you are looking at cells. That said, >>>>> John's >>>>> comments about shorter UV imaging are still valid; it's easier >>>>> to get a >>>>> >>>> NIR >>>> >>>>> Ti:Saph beam through conventional optics than a very UV beam. >>>>> This only >>>>> applies if you are using dyes that need very UV excitation of >>>>> course. >>>>> >>>>> Craig >>>>> >>>>> >>>>> >>>>> >>>>> On Sun, Apr 24, 2011 at 6:29 AM, George McNamara >>>>> <[hidden email]>wrote: >>>>> >>>>> ***** >>>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>>> ***** >>>>>> >>>>>> Hi David, >>>>>> >>>>>> The tuning range of Ti:Sapphire (ex. 690-1040 nm) used to be >>>>>> problematic >>>>>> for exciting far red and infrared fluorophores. There were some >>>>>> >>>>> exceptions, >>>> >>>>> such as Alexa Fluor 488m 568, 594, and 633 all exciting well >>>>> around 755 >>>>>> >>>>> nm >>>> >>>>> (Fig 5D of Dickinson et al 2003 J Biomed Optics 8: 329-338), but >>>>> since >>>>>> >>>>> you >>>> >>>>> are ok with cost and complexity, adding an OPO(s) and/or >>>>> multiple MP >>>>>> >>>>> lasers, >>>> >>>>> gives you full range (see >>>>>> http://www.coherent.com/products/?1570/Chameleon-OPO-Tuning- >>>>>> Range for >>>>>> OPO). >>>>>> >>>>>> Depending on cell type and physiological needs, you may be able >>>>>> to >>>>>> >>>>> suppress >>>> >>>>> phototoxicity by lowering O2 - cell culture at ~18% O2 is >>>>> something of >>>>>> >>>>> an >>>> >>>>> artifact for many cell types - by using catalase/glucose oxidase/ >>>>> glucose >>>>>> >>>>> or >>>> >>>>> Oxyrase (for the latter, www.oxyrase.com, see Waterman-Storer >>>>> 1993 and >>>>>> Mikhailov 1995 Cell Motil Cytokseleton). >>>>>> >>>>>> Enjoy, >>>>>> >>>>>> George >>>>>> >>>>>> >>>>>> On 4/23/2011 6:03 PM, David Knecht charter wrote: >>>>>> >>>>>> ***** >>>>>>> To join, leave or search the confocal microscopy listserv, go >>>>>>> to: >>>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>>>> ***** >>>>>>> >>>>>>> Is there any disadvantage to multiphoton for cultured cells >>>>>>> (besides >>>>>>> >>>>>> cost >>>> >>>>> and complexity)? Cell viability? Phototoxicity? Dave >>>>>>> >>>>>>> On Apr 23, 2011, at 11:21 AM, Kate Luby-Phelps wrote: >>>>>>> >>>>>>> >>>>>>> >>>>>>> ***** >>>>>>>> To join, leave or search the confocal microscopy listserv, go >>>>>>>> to: >>>>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>>>>> ***** >>>>>>>> >>>>>>>> Multiphoton has no clear advantage for cells in culture. For >>>>>>>> cell >>>>>>>> >>>>>>> culture >>>> >>>>> samples, >>>>>>>> we use two photon only to excite DAPI or other UV and near- >>>>>>>> UV fluors >>>>>>>> since we >>>>>>>> had to make a choice between the Ti:S and a 405 laser on our >>>>>>>> scope. >>>>>>>> >>>>>>>> Kate Luby-Phelps >>>>>>>> >>>>>>>> >>>>>>>> Dr. David Knecht >>>>>>> Department of Molecular and Cell Biology >>>>>>> Co-head Flow Cytometry and Confocal Microscopy Facility >>>>>>> U-3125 >>>>>>> 91 N. Eagleville Rd. >>>>>>> University of Connecticut >>>>>>> Storrs, CT 06269 >>>>>>> 860-486-2200 >>>>>>> 860-486-4331 (fax) >>>>>>> >>>>>>> >>>>>>> >>>>>>> >>>>>> >>>>>> -- >>>>>> >>>>>> >>>>>> George McNamara, PhD >>>>>> Analytical Imaging Core Facility >>>>>> University of Miami >>>>>> >>>>>> >>>> |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Ah, that makes sense to me now. Less intensity, but more fluorophores exposed to that intensity as the area increases, leading to a net linear response. Thanks! Craig On Thu, Apr 28, 2011 at 2:03 PM, Mark Cannell <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Yes but the area is increased by a similar amount so the total released > does not fall with distance (assuming no inner filtering effects etc) > > Cheers > > On 29/04/2011, at 7:41 AM, Craig Brideau wrote: > > ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> So if it is linear with intensity (1P), should it roll off to 1/4 in terms >> of distance from the ideal focal point? Intensity drops to 1/4 at 1 unit >> distance from a point source, so I'm approximating the center of the focal >> point here. Am I right in my thinking? >> >> Craig >> >> >> >> On Wed, Apr 27, 2011 at 8:38 PM, Mark Cannell <[hidden email] >> >wrote: >> >> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> At low powers the uncaging is linear in intensity so there should be no >>> threshold. So, the uncaging volume at low powers is the same as the PSF >>> and >>> for 2P its the 2P PSF. If ground state depletion develops the FWHM of the >>> focal uncaging spot grows (of course). >>> >>> Mark >>> >>> >>> On 28/04/2011, at 2:18 PM, Craig Brideau wrote: >>> >>> ***** >>> >>>> To join, leave or search the confocal microscopy listserv, go to: >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>> ***** >>>> >>>> How much uncaging occurs outside the focal region in single photon? If >>>> you >>>> don't have your laser turned up to much the energy density should only >>>> exceed the uncaging threshold near the focal point, yes? >>>> >>>> Craig >>>> >>>> >>>> On Wed, Apr 27, 2011 at 6:48 PM, Rosemary White <[hidden email] >>>> >>>>> wrote: >>>>> >>>> >>>> ***** >>>> >>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>> ***** >>>>> >>>>> Dear all, >>>>> >>>>> Haven't followed this discussion closely, but would one advantage of 2P >>>>> lie >>>>> in the ability to uncage caged compounds - fluorescent tracers, for >>>>> example, >>>>> within a single cell? We've used UV uncaging and while you get most >>>>> uncaging in the target cell, you also get cones of uncaging above and >>>>> below >>>>> the plane of focus. Seems that 2P would overcome this problem. >>>>> >>>>> Rosemary White >>>>> >>>>> Dr Rosemary White >>>>> CSIRO Plant Industry >>>>> GPO Box 1600 >>>>> Canberra, ACT 2601 >>>>> Australia >>>>> >>>>> T 61 2 6246 5475 >>>>> F 61 2 6246 5334 >>>>> E [hidden email] >>>>> >>>>> >>>>> >>>>> On 28/04/11 10:10 AM, "Craig Brideau" <[hidden email]> wrote: >>>>> >>>>> ***** >>>>> >>>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>>> ***** >>>>>> >>>>>> I second the previous lister's opinions. At the end of the day MP is >>>>>> >>>>>> really >>>>> >>>>> only good for thick tissue sections. If you are looking at cell >>>>>> layers >>>>>> >>>>>> just >>>>> >>>>> use a conventional 1P confocal. The key advantage of 2P is >>>>>> penetration >>>>>> depth, which is moot when you are looking at cells. That said, John's >>>>>> comments about shorter UV imaging are still valid; it's easier to get >>>>>> a >>>>>> >>>>>> NIR >>>>> >>>>> Ti:Saph beam through conventional optics than a very UV beam. This >>>>>> only >>>>>> applies if you are using dyes that need very UV excitation of course. >>>>>> >>>>>> Craig >>>>>> >>>>>> >>>>>> >>>>>> >>>>>> On Sun, Apr 24, 2011 at 6:29 AM, George McNamara >>>>>> <[hidden email]>wrote: >>>>>> >>>>>> ***** >>>>>> >>>>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>>>> ***** >>>>>>> >>>>>>> Hi David, >>>>>>> >>>>>>> The tuning range of Ti:Sapphire (ex. 690-1040 nm) used to be >>>>>>> problematic >>>>>>> for exciting far red and infrared fluorophores. There were some >>>>>>> >>>>>>> exceptions, >>>>>> >>>>> >>>>> such as Alexa Fluor 488m 568, 594, and 633 all exciting well around >>>>>> 755 >>>>>> >>>>>>> >>>>>>> nm >>>>>> >>>>> >>>>> (Fig 5D of Dickinson et al 2003 J Biomed Optics 8: 329-338), but since >>>>>> >>>>>>> >>>>>>> you >>>>>> >>>>> >>>>> are ok with cost and complexity, adding an OPO(s) and/or multiple MP >>>>>> >>>>>>> >>>>>>> lasers, >>>>>> >>>>> >>>>> gives you full range (see >>>>>> >>>>>>> http://www.coherent.com/products/?1570/Chameleon-OPO-Tuning-Rangefor >>>>>>> OPO). >>>>>>> >>>>>>> Depending on cell type and physiological needs, you may be able to >>>>>>> >>>>>>> suppress >>>>>> >>>>> >>>>> phototoxicity by lowering O2 - cell culture at ~18% O2 is something of >>>>>> >>>>>>> >>>>>>> an >>>>>> >>>>> >>>>> artifact for many cell types - by using catalase/glucose >>>>>> oxidase/glucose >>>>>> >>>>>>> >>>>>>> or >>>>>> >>>>> >>>>> Oxyrase (for the latter, www.oxyrase.com, see Waterman-Storer 1993 >>>>>> and >>>>>> >>>>>>> Mikhailov 1995 Cell Motil Cytokseleton). >>>>>>> >>>>>>> Enjoy, >>>>>>> >>>>>>> George >>>>>>> >>>>>>> >>>>>>> On 4/23/2011 6:03 PM, David Knecht charter wrote: >>>>>>> >>>>>>> ***** >>>>>>> >>>>>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>>>>> ***** >>>>>>>> >>>>>>>> Is there any disadvantage to multiphoton for cultured cells (besides >>>>>>>> >>>>>>>> cost >>>>>>> >>>>>> >>>>> and complexity)? Cell viability? Phototoxicity? Dave >>>>>> >>>>>>> >>>>>>>> On Apr 23, 2011, at 11:21 AM, Kate Luby-Phelps wrote: >>>>>>>> >>>>>>>> >>>>>>>> >>>>>>>> ***** >>>>>>>> >>>>>>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>>>>>> ***** >>>>>>>>> >>>>>>>>> Multiphoton has no clear advantage for cells in culture. For cell >>>>>>>>> >>>>>>>>> culture >>>>>>>> >>>>>>> >>>>> samples, >>>>>> >>>>>>> we use two photon only to excite DAPI or other UV and near-UV fluors >>>>>>>>> since we >>>>>>>>> had to make a choice between the Ti:S and a 405 laser on our scope. >>>>>>>>> >>>>>>>>> Kate Luby-Phelps >>>>>>>>> >>>>>>>>> >>>>>>>>> Dr. David Knecht >>>>>>>>> >>>>>>>> Department of Molecular and Cell Biology >>>>>>>> Co-head Flow Cytometry and Confocal Microscopy Facility >>>>>>>> U-3125 >>>>>>>> 91 N. Eagleville Rd. >>>>>>>> University of Connecticut >>>>>>>> Storrs, CT 06269 >>>>>>>> 860-486-2200 >>>>>>>> 860-486-4331 (fax) >>>>>>>> >>>>>>>> >>>>>>>> >>>>>>>> >>>>>>>> >>>>>>> -- >>>>>>> >>>>>>> >>>>>>> George McNamara, PhD >>>>>>> Analytical Imaging Core Facility >>>>>>> University of Miami >>>>>>> >>>>>>> >>>>>>> >>>>> |