|
Daniel Fimiarz |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hello, I am currently researching a possibility of acquiring a multi photon system for our imaging core facility. What are typical maintenance costs of such systems? Are they very different from single photon (confocal) systems? I do realize that my question is rather general and I am not expecting very detailed answers. Anything that you could share would be helpful. Thank you very much, Daniel F. |
|
Armstrong, Brian |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal For a general answer to your general question, I would say that the difference is the laser. A service contract is roughly 10% of the cost of the equipment so a service contract on a 2P laser will be more than a vis laser. A 2P laser is around $200k and the contract is around $15k yearly. Also, given that a 2P system would lack a scan head it may be more simple and less expensive to maintain. However, you may consider a service contract here as well which would add expense after the first year following normal warranty coverage. If you want to talk more about 2P systems you can contact me directly. Cheers, Brian D Armstrong PhD Light Microscopy Core Manager Beckman Research Institute City of Hope 1450 E Duarte Rd Duarte, CA 91010 626-359-8111 x62872 http://www.cityofhope.org/SharedResources/LightMicroscopy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Daniel Fimiarz Sent: Tuesday, October 09, 2007 7:06 AM To: [hidden email] Subject: Maintenance costs of multi photon systems Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hello, I am currently researching a possibility of acquiring a multi photon system for our imaging core facility. What are typical maintenance costs of such systems? Are they very different from single photon (confocal) systems? I do realize that my question is rather general and I am not expecting very detailed answers. Anything that you could share would be helpful. Thank you very much, Daniel F. "EMF <COH.org>" made the following annotations. ------------------------------------------------------------------------------ SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. ============================================================================== |
|
Craig Brideau |
|
|
In reply to this post by Daniel Fimiarz
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal In terms of the laser it self, there are a few things to consider. If you go with a Ti:Saph laser, it doesn't actually need a whole lot from day to day. From my experience with our own lab's laser, it needs an occasional top-up of distilled/filtered water for the cooling chiller (maybe a liter once every couple months!). If you plan on running the laser below 860nm or so you generally have to purge the laser cavity with ultra pure (5.0) dry nitrogen. Our own lab has a researcher that works at 925nm and we go through a tank every couple months. Finally, most Ti:Saphs use diode lasers as a pump source. These diodes can work for anywhere from 5 to 10 years, and cost about 10-20k to replace. From a facilities standpoint, Ti:Saphs require a fairly clean, temperature-stable environment. They put out a lot of heat from the chiller, so it is important to ensure your room has adequate air conditioning to handle the thermal load. If the temperature changes more than a couple degrees over the course of a day it can be bad for the laser. Hope this helps! Craig On 10/9/07, Daniel Fimiarz <[hidden email]> wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Hello, > > I am currently researching a possibility of acquiring a multi photon > system for our imaging core facility. What are typical maintenance > costs of such systems? Are they very different from single photon > (confocal) systems? I do realize that my question is rather general > and I am not expecting very detailed answers. Anything that you could > share would be helpful. > > Thank you very much, > > Daniel F. > |
|
Mark Cannell |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Craig Why use expensive ultra pure nitrogen? I thought it was mostly water vapor that you need to reduce in the cavity so that regular grade nitrogen should be OK. (the H2O absorbance reduces the cavity gain) Come to that, even dried air might be suitable. Cheers Mark On 10/10/2007, at 6:13 AM, Craig Brideau wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > In terms of the laser it self, there are a few things to consider. If > you go with a Ti:Saph laser, it doesn't actually need a whole lot from > day to day. From my experience with our own lab's laser, it needs an > occasional top-up of distilled/filtered water for the cooling chiller > (maybe a liter once every couple months!). If you plan on running the > laser below 860nm or so you generally have to purge the laser cavity > with ultra pure (5.0) dry nitrogen. Our own lab has a researcher that > works at 925nm and we go through a tank every couple months. Finally, > most Ti:Saphs use diode lasers as a pump source. These diodes can > work for anywhere from 5 to 10 years, and cost about 10-20k to > replace. > From a facilities standpoint, Ti:Saphs require a fairly clean, > temperature-stable environment. They put out a lot of heat from the > chiller, so it is important to ensure your room has adequate air > conditioning to handle the thermal load. If the temperature changes > more than a couple degrees over the course of a day it can be bad for > the laser. > > Hope this helps! > > Craig > > On 10/9/07, Daniel Fimiarz <[hidden email]> wrote: >> Search the CONFOCAL archive at >> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >> >> Hello, >> >> I am currently researching a possibility of acquiring a multi photon >> system for our imaging core facility. What are typical maintenance >> costs of such systems? Are they very different from single photon >> (confocal) systems? I do realize that my question is rather general >> and I am not expecting very detailed answers. Anything that you could >> share would be helpful. >> >> Thank you very much, >> >> Daniel F. >> |
|
Karl Garsha |
|
|
In reply to this post by Daniel Fimiarz
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello Daniel, You should consider the maintenance/expendables associated with the pump laser; the diodes will likely need to be replaced after a certain number of hours, this may be a cost you want to plan ahead for. Karl
Garsha
The
Microimaging Applications Group comprises
five
imaging leaders: Photometrics, Gatan,
Media Cybernetics, Daniel Fimiarz wrote: Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hello, I am currently researching a possibility of acquiring a multi photon system for our imaging core facility. What are typical maintenance costs of such systems? Are they very different from single photon (confocal) systems? I do realize that my question is rather general and I am not expecting very detailed answers. Anything that you could share would be helpful. Thank you very much, Daniel F. |
|
|
In reply to this post by Mark Cannell
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
EELS Application Specialist |
|
Craig Brideau |
|
|
In reply to this post by Mark Cannell
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Two words: Dust Particles. Any dirt that makes it into the optical cavity will shorten the life of the laser. The reason to use the pure nitrogen is that its very, very clean. When you are dealing with the kind of power levels that get slung around inside the optical cavity of a Ti:Saph, you don't want any contamination! If you can get super clean dry air then more power to you, but why risk a $100k laser for the sake of a few bucks of pure n2? Craig On 10/10/07, Mark Cannell <[hidden email]> wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Hi Craig > > Why use expensive ultra pure nitrogen? I thought it was mostly water > vapor that you need to reduce in the cavity so that regular grade > nitrogen should be OK. (the H2O absorbance reduces the cavity gain) > Come to that, even dried air might be suitable. > > Cheers Mark > > On 10/10/2007, at 6:13 AM, Craig Brideau wrote: > > > Search the CONFOCAL archive at > > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > > > In terms of the laser it self, there are a few things to consider. If > > you go with a Ti:Saph laser, it doesn't actually need a whole lot from > > day to day. From my experience with our own lab's laser, it needs an > > occasional top-up of distilled/filtered water for the cooling chiller > > (maybe a liter once every couple months!). If you plan on running the > > laser below 860nm or so you generally have to purge the laser cavity > > with ultra pure (5.0) dry nitrogen. Our own lab has a researcher that > > works at 925nm and we go through a tank every couple months. Finally, > > most Ti:Saphs use diode lasers as a pump source. These diodes can > > work for anywhere from 5 to 10 years, and cost about 10-20k to > > replace. > > From a facilities standpoint, Ti:Saphs require a fairly clean, > > temperature-stable environment. They put out a lot of heat from the > > chiller, so it is important to ensure your room has adequate air > > conditioning to handle the thermal load. If the temperature changes > > more than a couple degrees over the course of a day it can be bad for > > the laser. > > > > Hope this helps! > > > > Craig > > > > On 10/9/07, Daniel Fimiarz <[hidden email]> wrote: > >> Search the CONFOCAL archive at > >> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >> > >> Hello, > >> > >> I am currently researching a possibility of acquiring a multi photon > >> system for our imaging core facility. What are typical maintenance > >> costs of such systems? Are they very different from single photon > >> (confocal) systems? I do realize that my question is rather general > >> and I am not expecting very detailed answers. Anything that you could > >> share would be helpful. > >> > >> Thank you very much, > >> > >> Daniel F. > >> > |
|
Mark Cannell |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi Craig Interesting idea, but in 10 years of operating a Ti:S, I've not noticed any problem with dust flying around inside the cavity. (You can see room dust when working on the cavity with the covers off). In any case, as far as I know, commercial compressed gasses _are_ filtered. Here's an alternative: N2 blow off from a liquid N2 dewar - it is dry and squeeky clean... Cheers Mark On 11/10/2007, at 5:25 PM, Craig Brideau wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Two words: Dust Particles. > Any dirt that makes it into the optical cavity will shorten the life > of the laser. The reason to use the pure nitrogen is that its very, > very clean. When you are dealing with the kind of power levels that > get slung around inside the optical cavity of a Ti:Saph, you don't > want any contamination! If you can get super clean dry air then more > power to you, but why risk a $100k laser for the sake of a few bucks > of pure n2? > > Craig > > > On 10/10/07, Mark Cannell <[hidden email]> wrote: >> Search the CONFOCAL archive at >> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >> >> Hi Craig >> >> Why use expensive ultra pure nitrogen? I thought it was mostly water >> vapor that you need to reduce in the cavity so that regular grade >> nitrogen should be OK. (the H2O absorbance reduces the cavity gain) >> Come to that, even dried air might be suitable. >> >> Cheers Mark >> >> On 10/10/2007, at 6:13 AM, Craig Brideau wrote: >> >>> Search the CONFOCAL archive at >>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >>> >>> In terms of the laser it self, there are a few things to >>> consider. If >>> you go with a Ti:Saph laser, it doesn't actually need a whole lot >>> from >>> day to day. From my experience with our own lab's laser, it >>> needs an >>> occasional top-up of distilled/filtered water for the cooling >>> chiller >>> (maybe a liter once every couple months!). If you plan on >>> running the >>> laser below 860nm or so you generally have to purge the laser cavity >>> with ultra pure (5.0) dry nitrogen. Our own lab has a researcher >>> that >>> works at 925nm and we go through a tank every couple months. >>> Finally, >>> most Ti:Saphs use diode lasers as a pump source. These diodes can >>> work for anywhere from 5 to 10 years, and cost about 10-20k to >>> replace. >>> From a facilities standpoint, Ti:Saphs require a fairly clean, >>> temperature-stable environment. They put out a lot of heat from the >>> chiller, so it is important to ensure your room has adequate air >>> conditioning to handle the thermal load. If the temperature changes >>> more than a couple degrees over the course of a day it can be bad >>> for >>> the laser. >>> >>> Hope this helps! >>> >>> Craig >>> >>> On 10/9/07, Daniel Fimiarz <[hidden email]> wrote: >>>> Search the CONFOCAL archive at >>>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >>>> >>>> Hello, >>>> >>>> I am currently researching a possibility of acquiring a multi >>>> photon >>>> system for our imaging core facility. What are typical maintenance >>>> costs of such systems? Are they very different from single photon >>>> (confocal) systems? I do realize that my question is rather >>>> general >>>> and I am not expecting very detailed answers. Anything that you >>>> could >>>> share would be helpful. >>>> >>>> Thank you very much, >>>> >>>> Daniel F. >>>> >> |
|
Craig Brideau |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Our lab is very dusty due to an old ventilation system, so our problem may be worse than yours. As with any lab and laser system, "your mileage may vary" depending on your specific condtions. The N2 dewar idea is very interesting. What do you use in your system? Craig On 10/10/07, Mark Cannell <[hidden email]> wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Hi Craig > > Interesting idea, but in 10 years of operating a Ti:S, I've not > noticed any problem with dust flying around inside the cavity. (You > can see room dust when working on the cavity with the covers off). In > any case, as far as I know, commercial compressed gasses _are_ > filtered. Here's an alternative: N2 blow off from a liquid N2 dewar - > it is dry and squeeky clean... > > Cheers Mark > > > On 11/10/2007, at 5:25 PM, Craig Brideau wrote: > > > Search the CONFOCAL archive at > > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > > > Two words: Dust Particles. > > Any dirt that makes it into the optical cavity will shorten the life > > of the laser. The reason to use the pure nitrogen is that its very, > > very clean. When you are dealing with the kind of power levels that > > get slung around inside the optical cavity of a Ti:Saph, you don't > > want any contamination! If you can get super clean dry air then more > > power to you, but why risk a $100k laser for the sake of a few bucks > > of pure n2? > > > > Craig > > > > > > On 10/10/07, Mark Cannell <[hidden email]> wrote: > >> Search the CONFOCAL archive at > >> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >> > >> Hi Craig > >> > >> Why use expensive ultra pure nitrogen? I thought it was mostly water > >> vapor that you need to reduce in the cavity so that regular grade > >> nitrogen should be OK. (the H2O absorbance reduces the cavity gain) > >> Come to that, even dried air might be suitable. > >> > >> Cheers Mark > >> > >> On 10/10/2007, at 6:13 AM, Craig Brideau wrote: > >> > >>> Search the CONFOCAL archive at > >>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >>> > >>> In terms of the laser it self, there are a few things to > >>> consider. If > >>> you go with a Ti:Saph laser, it doesn't actually need a whole lot > >>> from > >>> day to day. From my experience with our own lab's laser, it > >>> needs an > >>> occasional top-up of distilled/filtered water for the cooling > >>> chiller > >>> (maybe a liter once every couple months!). If you plan on > >>> running the > >>> laser below 860nm or so you generally have to purge the laser cavity > >>> with ultra pure (5.0) dry nitrogen. Our own lab has a researcher > >>> that > >>> works at 925nm and we go through a tank every couple months. > >>> Finally, > >>> most Ti:Saphs use diode lasers as a pump source. These diodes can > >>> work for anywhere from 5 to 10 years, and cost about 10-20k to > >>> replace. > >>> From a facilities standpoint, Ti:Saphs require a fairly clean, > >>> temperature-stable environment. They put out a lot of heat from the > >>> chiller, so it is important to ensure your room has adequate air > >>> conditioning to handle the thermal load. If the temperature changes > >>> more than a couple degrees over the course of a day it can be bad > >>> for > >>> the laser. > >>> > >>> Hope this helps! > >>> > >>> Craig > >>> > >>> On 10/9/07, Daniel Fimiarz <[hidden email]> wrote: > >>>> Search the CONFOCAL archive at > >>>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > >>>> > >>>> Hello, > >>>> > >>>> I am currently researching a possibility of acquiring a multi > >>>> photon > >>>> system for our imaging core facility. What are typical maintenance > >>>> costs of such systems? Are they very different from single photon > >>>> (confocal) systems? I do realize that my question is rather > >>>> general > >>>> and I am not expecting very detailed answers. Anything that you > >>>> could > >>>> share would be helpful. > >>>> > >>>> Thank you very much, > >>>> > >>>> Daniel F. > >>>> > >> > |
|
Mark Cannell |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Hi We've only used junk grade N2 at 900 nm -but then we don't go there generally. Cheers On 11/10/2007, at 6:05 PM, Craig Brideau wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal > > Our lab is very dusty due to an old ventilation system, so our problem > may be worse than yours. As with any lab and laser system, "your > mileage may vary" depending on your specific condtions. The N2 dewar > idea is very interesting. What do you use in your system? > > Craig > > > On 10/10/07, Mark Cannell <[hidden email]> wrote: >> Search the CONFOCAL archive at >> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >> >> Hi Craig >> >> Interesting idea, but in 10 years of operating a Ti:S, I've not >> noticed any problem with dust flying around inside the cavity. (You >> can see room dust when working on the cavity with the covers off). In >> any case, as far as I know, commercial compressed gasses _are_ >> filtered. Here's an alternative: N2 blow off from a liquid N2 dewar - >> it is dry and squeeky clean... >> >> Cheers Mark >> >> >> On 11/10/2007, at 5:25 PM, Craig Brideau wrote: >> >>> Search the CONFOCAL archive at >>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >>> >>> Two words: Dust Particles. >>> Any dirt that makes it into the optical cavity will shorten the life >>> of the laser. The reason to use the pure nitrogen is that its very, >>> very clean. When you are dealing with the kind of power levels that >>> get slung around inside the optical cavity of a Ti:Saph, you don't >>> want any contamination! If you can get super clean dry air then >>> more >>> power to you, but why risk a $100k laser for the sake of a few bucks >>> of pure n2? >>> >>> Craig >>> >>> >>> On 10/10/07, Mark Cannell <[hidden email]> wrote: >>>> Search the CONFOCAL archive at >>>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >>>> >>>> Hi Craig >>>> >>>> Why use expensive ultra pure nitrogen? I thought it was mostly >>>> water >>>> vapor that you need to reduce in the cavity so that regular grade >>>> nitrogen should be OK. (the H2O absorbance reduces the cavity gain) >>>> Come to that, even dried air might be suitable. >>>> >>>> Cheers Mark >>>> >>>> On 10/10/2007, at 6:13 AM, Craig Brideau wrote: >>>> >>>>> Search the CONFOCAL archive at >>>>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >>>>> >>>>> In terms of the laser it self, there are a few things to >>>>> consider. If >>>>> you go with a Ti:Saph laser, it doesn't actually need a whole lot >>>>> from >>>>> day to day. From my experience with our own lab's laser, it >>>>> needs an >>>>> occasional top-up of distilled/filtered water for the cooling >>>>> chiller >>>>> (maybe a liter once every couple months!). If you plan on >>>>> running the >>>>> laser below 860nm or so you generally have to purge the laser >>>>> cavity >>>>> with ultra pure (5.0) dry nitrogen. Our own lab has a researcher >>>>> that >>>>> works at 925nm and we go through a tank every couple months. >>>>> Finally, >>>>> most Ti:Saphs use diode lasers as a pump source. These diodes can >>>>> work for anywhere from 5 to 10 years, and cost about 10-20k to >>>>> replace. >>>>> From a facilities standpoint, Ti:Saphs require a fairly clean, >>>>> temperature-stable environment. They put out a lot of heat >>>>> from the >>>>> chiller, so it is important to ensure your room has adequate air >>>>> conditioning to handle the thermal load. If the temperature >>>>> changes >>>>> more than a couple degrees over the course of a day it can be bad >>>>> for >>>>> the laser. >>>>> >>>>> Hope this helps! >>>>> >>>>> Craig >>>>> >>>>> On 10/9/07, Daniel Fimiarz <[hidden email]> wrote: >>>>>> Search the CONFOCAL archive at >>>>>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal >>>>>> >>>>>> Hello, >>>>>> >>>>>> I am currently researching a possibility of acquiring a multi >>>>>> photon >>>>>> system for our imaging core facility. What are typical >>>>>> maintenance >>>>>> costs of such systems? Are they very different from single photon >>>>>> (confocal) systems? I do realize that my question is rather >>>>>> general >>>>>> and I am not expecting very detailed answers. Anything that you >>>>>> could >>>>>> share would be helpful. >>>>>> >>>>>> Thank you very much, >>>>>> >>>>>> Daniel F. >>>>>> >>>> >> |
|
|
In reply to this post by John Hyun-2
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal John, I noticed with interest your recent employment note on the Microscopy listserver and hope that you are having good success. Our publication, Microscopy Today, is mailed six times per year to over 16,000 microscopists -- each of whom has either specifically requested the publication or were automatically subscribed as an MSA member benefit. Our next issue closes soon and will be mailed on or about January 15th. You may consider placing your employment ad with us. We offer employment advertisements at the reduced rate of $475 for a quarter-page (3 5/8 in. W X 5 in H) insertion. Larger (3 5/8" W) employment ads can be purchased for an additional $90 per inch over 5 inches high. Should you be interested, normal procedure is for you to supply your text by email , or fax the copy to us at (727) 507-7102. We will typeset, add a border, etc., and email back a PDF draft copy of the advertisement with a quote at no obligation to you. After your price approval, proof, changes, corrections, and billing instructions, are received, we will publish. You may include a logo if you wish. Kindly advise rather soon should you be interested in this proposal as the timing for this issue is becoming critical. Regards, Ron Anderson, Editor Microscopy Today John Hyun wrote: > Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Please see job > openings for EM application professionals for Gatan, Inc below. Thank you. > > > EELS Application Specialist > > > Job responsibility: > Support Gatan’s highly successful Analytical TEM product line (GIF and > EELS spectrometers). > Typical responsibilities include working with key users to develop new > EELS/GIF applications, providing customer training and support, > conducting product demonstrations, representing Gatan at scientific > conferences, and assisting new product development. > > Certain amount of travel is required. > > Qualification: > > * Strong EELS background. > * Hands-on experience in EELS, EDS, STEM, and EFTEM. > * Prior experience in Gatan products (GIF or EELS spectrometers) > * A Ph.D. degree in physics, chemistry, materials science or > engineering. > * Publications of analytical TEM work (EELS or EFTEM) > * Excellent verbal, written, and interpersonal communication > skills in English > > > > > The position is based in Pleasanton, California and carries a salary > commensurate with experience and an attractive benefits package. > > Please submit your resume to [hidden email]. > > > Marketing Product Manager - Analytical TEM Products > > > Job responsibility: > Develop business strategies and product line roadmap, acquire direct > input from customers on new products, provide technical support for > customers and sales, conduct workshops and seminars. Responsible for > product line P&L. > > Certain amount of travel is required. > > Qualification: > > * Previous commercial experience in product management for a > scientific instrument company (electron microscopy industry) is > preferred but not necessary > * Knowledge and experience in TEM/STEM, EELS, EDS, and EFTEM. > * Previous experience in Gatan products (GIF or EELS > spectrometers) preferred > * A Ph.D. degree in physics, chemistry, materials science or > engineering. > * Excellent verbal, written, and interpersonal communication > skills in English > > > > The position is based in Pleasanton, California and carries a salary > commensurate with experience and an attractive benefits package. > > Please submit your resume to [hidden email] > > > > 3View� Application Specialist > > > 3View is Gatan’s latest product aimed at acquiring 3D image data in a > SEM from life science specimens over significant volumes with large > field of view and high resolution. 3View is based on the ground > breaking technique developed by MPI Heidelberg (W Denk and H > Horstmann, Max-Planck Institute for Medical Research, Heidelberg, > Germany. /“Serial block-face scanning electron microscopy to > reconstruct three-dimensional tissue nanostructure”/. Plos Biology, > 2004.2(11):p. 1900-1909), with special application to the neuroscience > community > > Job responsibility: > Provide technical support of 3View product. Responsibilities include > product demonstrations, presentations, working with customers to > acquire practical data, generation of sales and marketing literature, > and definition of new product features. > Frequent worldwide travel may be required. > > Qualification: > > * Hands-on experience in sample preparations for ultramicrotomy > (fixation and embedment, and subsequent generation of ultra-thin > sections > * Hands-on experience of SEM/TEM and/or light microscopy (LM) > * An advanced degree in related life science field is preferred > * Experience using imaging software including 3D image reconstruction > * Excellent verbal, written, and interpersonal communication > skills in English > > > > The position is based in Pleasanton, California and carries a salary > commensurate with experience and an attractive benefits package. > > Please submit your resume to [hidden email] > <mailto:[hidden email]> and mention 3View Application > Specialist. |
| Free forum by Nabble | Edit this page |