To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Colleagues, Could someone suggest me a fluorophore with lifetime short enough to measure IRF in red channel (excitation: 520nm/ Detection : 565 – 605)? I would be very grateful for an advice. Best regards, Hanna -- Hanna Sas-Nowosielska Laboratory of Imaging Tissue Structure and Function Nencki Institute of Experimental Biology Warsaw, Poland |
![]() ![]() |
Joao Lagarto |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Hanna, Erythrosin B in water (tau = 90 ps, refs. Talbot et al (2015) Journal of Fluorescence & Boens et al (2007) Anal. Chem) should give you a nice IRF. Cheers, João João Lagarto, PhD Instituto Gulbenkian de Ciência Rua da Quinta Grande, 6 2780-156 Oeiras Portugal Às 12:49 de 19/11/2015, Hanna Sas Nowosielska escreveu: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear Colleagues, > > > > Could someone suggest me a fluorophore with lifetime short enough to > measure IRF in red channel (excitation: 520nm/ Detection : 565 – 605)? I > would be very grateful for an advice. > > > > Best regards, > > Hanna > > > |
![]() ![]() |
Kurt Anderson-3 |
![]() |
In reply to this post by Hanna_SN
try gold nano rods, as described here: Application of ultrafast gold luminescence to measuring the instrument response function for multispectral multiphoton fluorescence lifetime imaging. Optics Express: 2011 Jul 18;19(15):13848-61. doi: 10.1364/OE.19.013848. Prof. Kurt I. Anderson Tumor Cell Migration Lab and Beatson Advanced Imaging Resource (BAIR) The Beatson Institute for Cancer Research Garscube Estate, Switchback Road, Glasgow G61 1BD • Direct Line +44 (0) 141 330 2864 • Fax +44 (0) 141 942 6521 [hidden email]<https://mail.campus.gla.ac.uk/owa/redir.aspx?SURL=6BUWCm6RiBS6ZKio0WlxIllBRtFDaW3Hd6uHNmw7VVP73rNTf9jSCG0AYQBpAGwAdABvADoAawAuAGEAbgBkAGUAcgBzAG8AbgBAAGIAZQBhAHQAcwBvAG4ALgBnAGwAYQAuAGEAYwAuAHUAawA.&URL=mailto%3ak.anderson%40beatson.gla.ac.uk> A company limited by guarantee; Registered in Scotland No 84170; A Registered Scottish Charity No SC006106 The information contained in this communication may be privileged and confidential and is intended for the exclusive use of the addressee designated above. If you are not the addressee, any disclosure, reproduction, distribution or other dissemination or use of this communication is strictly prohibited. If you have received this transmission in error please would you be kind enough to inform us. On 19 Nov 2015, at 12:49, Hanna Sas Nowosielska <[hidden email]<mailto:[hidden email]>> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Colleagues, Could someone suggest me a fluorophore with lifetime short enough to measure IRF in red channel (excitation: 520nm/ Detection : 565 – 605)? I would be very grateful for an advice. Best regards, Hanna -- Hanna Sas-Nowosielska Laboratory of Imaging Tissue Structure and Function Nencki Institute of Experimental Biology Warsaw, Poland |
![]() ![]() |
Peter Kapusta |
![]() |
In reply to this post by Hanna_SN
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Erythrosin B's lifetime is still too long, but you can make it shorter by saturating the solution by KI (Potassium Iodide): http://dx.doi.org/10.1366/000370209787598979 The same trick works with Rose Bengal as well: http://www.sciencedirect.com/science/article/pii/S0009261409001389 The dye LDS 798 is another alternative http://scitation.aip.org/content/aip/journal/rsi/80/3/10.1063/1.3095677 Best regards, Peter |
![]() ![]() |
yuansheng sun |
![]() |
In reply to this post by Hanna_SN
you can try to use the cream coffee mate solution to record the laser scatter directly.
Sheng Sent from my T-Mobile 4G LTE Device -------- Original message -------- From: Hanna Sas Nowosielska <[hidden email]> Date:11/19/2015 6:49 AM (GMT-06:00) To: [hidden email] Cc: Subject: Measuring IRF in red channel ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Colleagues, Could someone suggest me a fluorophore with lifetime short enough to measure IRF in red channel (excitation: 520nm/ Detection : 565 – 605)? I would be very grateful for an advice. Best regards, Hanna -- Hanna Sas-Nowosielska Laboratory of Imaging Tissue Structure and Function Nencki Institute of Experimental Biology Warsaw, Poland |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thank you for all suggestions, unfortunately we cannot use scatter light to get IRF. There is some problem either with our WLL laser or with our microscope system (Leica SP8) and when using scatter light we don’t obtain specific IRF. Instead we get multiple picks of different height and size. We talked to service but they’re still trying to find the source of the problem. Hanna 2015-11-20 3:25 GMT+01:00 yuansheng.sun <[hidden email]>: > you can try to use the cream coffee mate solution to record the laser > scatter directly. > > Sheng > > > Sent from my T-Mobile 4G LTE Device > > > -------- Original message -------- > From: Hanna Sas Nowosielska <[hidden email]> > Date:11/19/2015 6:49 AM (GMT-06:00) > To: [hidden email] > Cc: > Subject: Measuring IRF in red channel > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear Colleagues, > > > > Could someone suggest me a fluorophore with lifetime short enough to > measure IRF in red channel (excitation: 520nm/ Detection : 565 – 605)? I > would be very grateful for an advice. > > > > Best regards, > > Hanna > > > > -- > Hanna Sas-Nowosielska > > Laboratory of Imaging Tissue Structure and Function > Nencki Institute of Experimental Biology > Warsaw, Poland > -- Hanna Sas-Nowosielska Laboratory of Imaging Tissue Structure and Function Nencki Institute of Experimental Biology Warsaw, Poland |
![]() ![]() |
Vladimir Ghukasyan-2 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Hanna, There may be several reasons to the problem depending on the pattern. It may be "ringing" or reflections. While the first one can be fixed by adjusting the threshold of the signal (increasing the baseline and getting rid of some noise in the microscope environment), the second problem will be more difficult to address in a turn-key system. This may be coming from two reflecting surfaces (mirrors or filters). It is most probably in the laser coupling part of the setup. This pattern will have to be addressed anyway before you can do any measurements. Best wishes, Vladimir On Fri, Nov 20, 2015 at 3:14 AM, Hanna Sas Nowosielska < [hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Thank you for all suggestions, unfortunately we cannot use scatter light to > get IRF. There is some problem either with our WLL laser or with our > microscope system (Leica SP8) and when using scatter light we don’t obtain > specific IRF. Instead we get multiple picks of different height and size. > We talked to service but they’re still trying to find the source of the > problem. > > > Hanna > > 2015-11-20 3:25 GMT+01:00 yuansheng.sun <[hidden email]>: > > > you can try to use the cream coffee mate solution to record the laser > > scatter directly. > > > > Sheng > > > > > > Sent from my T-Mobile 4G LTE Device > > > > > > -------- Original message -------- > > From: Hanna Sas Nowosielska <[hidden email]> > > Date:11/19/2015 6:49 AM (GMT-06:00) > > To: [hidden email] > > Cc: > > Subject: Measuring IRF in red channel > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Dear Colleagues, > > > > > > > > Could someone suggest me a fluorophore with lifetime short enough to > > measure IRF in red channel (excitation: 520nm/ Detection : 565 – 605)? I > > would be very grateful for an advice. > > > > > > > > Best regards, > > > > Hanna > > > > > > > > -- > > Hanna Sas-Nowosielska > > > > Laboratory of Imaging Tissue Structure and Function > > Nencki Institute of Experimental Biology > > Warsaw, Poland > > > > > > -- > Hanna Sas-Nowosielska > > Laboratory of Imaging Tissue Structure and Function > Nencki Institute of Experimental Biology > Warsaw, Poland > |
![]() ![]() |
yuansheng sun |
![]() |
In reply to this post by Hanna_SN
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Hanna, Since you use WLL on SP8, did you play around the AOBM (excitation selected from WLL) and the AOBS (emission window) to see how the IRFs change at different excitation/emission wavelengths? This may tell you something for what is going on. Good luck with your measurements. Sheng On Fri, Nov 20, 2015 at 2:14 AM, Hanna Sas Nowosielska < [hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Thank you for all suggestions, unfortunately we cannot use scatter light to > get IRF. There is some problem either with our WLL laser or with our > microscope system (Leica SP8) and when using scatter light we don’t obtain > specific IRF. Instead we get multiple picks of different height and size. > We talked to service but they’re still trying to find the source of the > problem. > > > Hanna > > 2015-11-20 3:25 GMT+01:00 yuansheng.sun <[hidden email]>: > > > you can try to use the cream coffee mate solution to record the laser > > scatter directly. > > > > Sheng > > > > > > Sent from my T-Mobile 4G LTE Device > > > > > > -------- Original message -------- > > From: Hanna Sas Nowosielska <[hidden email]> > > Date:11/19/2015 6:49 AM (GMT-06:00) > > To: [hidden email] > > Cc: > > Subject: Measuring IRF in red channel > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Dear Colleagues, > > > > > > > > Could someone suggest me a fluorophore with lifetime short enough to > > measure IRF in red channel (excitation: 520nm/ Detection : 565 – 605)? I > > would be very grateful for an advice. > > > > > > > > Best regards, > > > > Hanna > > > > > > > > -- > > Hanna Sas-Nowosielska > > > > Laboratory of Imaging Tissue Structure and Function > > Nencki Institute of Experimental Biology > > Warsaw, Poland > > > > > > -- > Hanna Sas-Nowosielska > > Laboratory of Imaging Tissue Structure and Function > Nencki Institute of Experimental Biology > Warsaw, Poland > |