|
Holly L. AARON |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear All - We have been struggling (for some time) with a Nikon/Andor spinning disk system running Metamorph and I am now wondering about trying MicroManager on this system. It is in regular use, so we cannot afford to disrupt the users who are not having issues. One major problem right now is with the PFS and a difficult sample with a matrix gel and multiple well-plates. Anyway, my question to the experts out there is whether or not we can run uManager simultaneously on a system running Metamorph or if they will interfere with each other and we should not bother? If they won't play well together, I would be interested to know if there is a Metamorph User Community out there? Please point me the way! We are not able to get any support from any of our vendors on this software and feel a bit stuck with it now. Thanks! -Holly __________________________________________________________________ Holly L. Aaron Molecular Imaging Center, Cancer Research Laboratory University of California, Berkeley 251 Life Sciences Addition, MC #2751 Berkeley, CA 94720-2751 510.642.2901 office 510.426.5005 GoogleVoice [hidden email] http://imaging.berkeley.edu Also: 444-B Li Ka Shing Center for Biomedical and Health Sciences |
|
Kurt Thorn |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Holly - There should be no problem with running Metamorph and Micromanager simultaneously. With very rare exceptions, Micro-manager will run with the standard drivers provided with the manufacturer. Even if it does require a special driver, this would have to be installed separately, and so installing Micro-manager won't touch any of the drivers you are currently using. We have a Nikon spinning disk system running both NIS-Elements and Micro-manager, and we've run several Andor systems under Micro-manager as well, so I don't think this should be a problem. Feel free to contact me directly if you want more advice or help with getting this set up. Best, Kurt On 9/30/2015 10:35 AM, Holly L. AARON wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear All - > > We have been struggling (for some time) with a Nikon/Andor spinning disk > system running Metamorph and I am now wondering about trying MicroManager > on this system. It is in regular use, so we cannot afford to disrupt the > users who are not having issues. One major problem right now is with the > PFS and a difficult sample with a matrix gel and multiple well-plates. > > Anyway, my question to the experts out there is whether or not we can run > uManager simultaneously on a system running Metamorph or if they will > interfere with each other and we should not bother? > > If they won't play well together, I would be interested to know if there is > a Metamorph User Community out there? Please point me the way! We are not > able to get any support from any of our vendors on this software and feel a > bit stuck with it now. > > Thanks! > -Holly > __________________________________________________________________ > Holly L. Aaron > Molecular Imaging Center, Cancer Research Laboratory > University of California, Berkeley > 251 Life Sciences Addition, MC #2751 > Berkeley, CA 94720-2751 > 510.642.2901 office 510.426.5005 GoogleVoice > [hidden email] http://imaging.berkeley.edu > Also: > 444-B Li Ka Shing Center for Biomedical and Health Sciences > -- Kurt Thorn Associate Professor Director, Nikon Imaging Center http://thornlab.ucsf.edu/ http://nic.ucsf.edu/blog/ |
|
Csúcs Gábor-3 |
|
|
In reply to this post by Holly L. AARON
Dear Holly,
Though we didn't do this with a spinning disk system (just a fully motorized wide-field system) we didn't observe any issues having Metamorph and MicoManager on the same machine. Of course you may want to make sure that no one tries to run them at the same time as this may cause trouble. Alternatively, you could set up the system as a dual-boot one and in this case you should be able to avoid any interference. Greetings Gabor -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Holly L. AARON Sent: Wednesday, September 30, 2015 7:36 PM To: [hidden email] Subject: Micromanager compatibility with Metamorph? ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear All - We have been struggling (for some time) with a Nikon/Andor spinning disk system running Metamorph and I am now wondering about trying MicroManager on this system. It is in regular use, so we cannot afford to disrupt the users who are not having issues. One major problem right now is with the PFS and a difficult sample with a matrix gel and multiple well-plates. Anyway, my question to the experts out there is whether or not we can run uManager simultaneously on a system running Metamorph or if they will interfere with each other and we should not bother? If they won't play well together, I would be interested to know if there is a Metamorph User Community out there? Please point me the way! We are not able to get any support from any of our vendors on this software and feel a bit stuck with it now. Thanks! -Holly __________________________________________________________________ Holly L. Aaron Molecular Imaging Center, Cancer Research Laboratory University of California, Berkeley 251 Life Sciences Addition, MC #2751 Berkeley, CA 94720-2751 510.642.2901 office 510.426.5005 GoogleVoice [hidden email] http://imaging.berkeley.edu Also: 444-B Li Ka Shing Center for Biomedical and Health Sciences |
|
Nico Stuurman-3 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Holly, Gabor is absolutely correct. There is no harm in installing Micro-Manager side by side with Metamorph. The only problem I can imagine is that one of the two would want a different driver version for a certain piece of hardware . The easiest way forward is to install Micro-Manager (which will not install drivers by itself) and see if it works with your hardware,. Nevertheless, it is possible that the issues your users with the Perfect Focus are related to the hardware/sample combination rather than the software. Best, Nico On 9/30/15 11:44 AM, Csúcs Gábor wrote: > Dear Holly, > > Though we didn't do this with a spinning disk system (just a fully motorized wide-field system) we didn't observe any issues having Metamorph and MicoManager on the same machine. Of course you may want to make sure that no one tries to run them at the same time as this may cause trouble. Alternatively, you could set up the system as a dual-boot one and in this case you should be able to avoid any interference. > > Greetings Gabor > > > > > > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Holly L. AARON > Sent: Wednesday, September 30, 2015 7:36 PM > To: [hidden email] > Subject: Micromanager compatibility with Metamorph? > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear All - > > We have been struggling (for some time) with a Nikon/Andor spinning disk system running Metamorph and I am now wondering about trying MicroManager on this system. It is in regular use, so we cannot afford to disrupt the users who are not having issues. One major problem right now is with the PFS and a difficult sample with a matrix gel and multiple well-plates. > > Anyway, my question to the experts out there is whether or not we can run uManager simultaneously on a system running Metamorph or if they will interfere with each other and we should not bother? > > If they won't play well together, I would be interested to know if there is a Metamorph User Community out there? Please point me the way! We are not able to get any support from any of our vendors on this software and feel a bit stuck with it now. > > Thanks! > -Holly > __________________________________________________________________ > Holly L. Aaron > Molecular Imaging Center, Cancer Research Laboratory University of California, Berkeley > 251 Life Sciences Addition, MC #2751 > Berkeley, CA 94720-2751 > 510.642.2901 office 510.426.5005 GoogleVoice > [hidden email] http://imaging.berkeley.edu > Also: > 444-B Li Ka Shing Center for Biomedical and Health Sciences -- Nico Stuurman Vale lab, UCSF/HHMI Genentech Hall, N316, MC2200 600 - 16th Street San Francisco, CA 94158-2517 415 514 3927 [hidden email] |
|
G. Esteban Fernandez |
|
|
In reply to this post by Kurt Thorn
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** For what it's worth, I have a Leica widefield and a Zeiss with 3rd party spinning disk that I run with Micro-Manager or manufacturer software installed on the same system. I have never experienced interference unless I tried to run both programs simultaneously. -Esteban On Sep 30, 2015 11:56 AM, "Kurt Thorn" <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Holly - > > There should be no problem with running Metamorph and Micromanager > simultaneously. With very rare exceptions, Micro-manager will run with the > standard drivers provided with the manufacturer. Even if it does require a > special driver, this would have to be installed separately, and so > installing Micro-manager won't touch any of the drivers you are currently > using. > > We have a Nikon spinning disk system running both NIS-Elements and > Micro-manager, and we've run several Andor systems under Micro-manager as > well, so I don't think this should be a problem. > > Feel free to contact me directly if you want more advice or help with > getting this set up. > > Best, > Kurt > > On 9/30/2015 10:35 AM, Holly L. AARON wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Dear All - >> >> We have been struggling (for some time) with a Nikon/Andor spinning disk >> system running Metamorph and I am now wondering about trying MicroManager >> on this system. It is in regular use, so we cannot afford to disrupt the >> users who are not having issues. One major problem right now is with the >> PFS and a difficult sample with a matrix gel and multiple well-plates. >> >> Anyway, my question to the experts out there is whether or not we can run >> uManager simultaneously on a system running Metamorph or if they will >> interfere with each other and we should not bother? >> >> If they won't play well together, I would be interested to know if there >> is >> a Metamorph User Community out there? Please point me the way! We are not >> able to get any support from any of our vendors on this software and feel >> a >> bit stuck with it now. >> >> Thanks! >> -Holly >> __________________________________________________________________ >> Holly L. Aaron >> Molecular Imaging Center, Cancer Research Laboratory >> University of California, Berkeley >> 251 Life Sciences Addition, MC #2751 >> Berkeley, CA 94720-2751 >> 510.642.2901 office 510.426.5005 GoogleVoice >> [hidden email] http://imaging.berkeley.edu >> Also: >> 444-B Li Ka Shing Center for Biomedical and Health Sciences >> >> > > -- > Kurt Thorn > Associate Professor > Director, Nikon Imaging Center > http://thornlab.ucsf.edu/ > http://nic.ucsf.edu/blog/ > |
|
Pariksheet Nanda |
|
|
In reply to this post by Holly L. AARON
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Holly, On Wed, Sep 30, 2015 at 5:35 PM, Holly L. AARON <[hidden email]> wrote: > I would be interested to know if there is > a Metamorph User Community out there? Please point me the way! We are not > able to get any support from any of our vendors on this software and feel a > bit stuck with it now. Since no one has mentioned it yet, I just thought I'd point out that, starting today, commercial support for Micro-Manager is available directly from the developers through Open-Imaging (https://open-imaging.com/). Additionally, Micro-Manager has an excellent public community through the mailing list (https://micro-manager.org/wiki/Micro-Manager%20Community). > -Holly Pariksheet |
|
WAINWRIGHT James |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** *** COMMERCIAL RESPONSE *** Hi Holly, I'm a microscopy support engineer for Andor in the UK and have some (limited) experience running our microscopy systems under uManager. Please feel free to contact me off-list if you need any technical information on setting up any of your hardware supplied from Andor. Best wishes! James James Wainwright Product Support Engineer (Microscopy Systems) Europe, Middle East, Africa Tel: +44 (0) 2890 270 873 Mob: +44 (0) 7834 710 834 Fax: +44 (0) 2890 310 792 Web: andor.com/support ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of Pariksheet Nanda [[hidden email]] Sent: 02 October 2015 03:42 To: [hidden email] Subject: Re: Micromanager compatibility with Metamorph? ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Holly, On Wed, Sep 30, 2015 at 5:35 PM, Holly L. AARON <[hidden email]> wrote: > I would be interested to know if there is > a Metamorph User Community out there? Please point me the way! We are not > able to get any support from any of our vendors on this software and feel a > bit stuck with it now. Since no one has mentioned it yet, I just thought I'd point out that, starting today, commercial support for Micro-Manager is available directly from the developers through Open-Imaging (https://open-imaging.com/). Additionally, Micro-Manager has an excellent public community through the mailing list (https://micro-manager.org/wiki/Micro-Manager%20Community). > -Holly Pariksheet This message has been scanned for malware by SurfControl plc. www.surfcontrol.com |
| Free forum by Nabble | Edit this page |