Microscope Air Tables

classic Classic list List threaded Threaded
♦
♦
9 messages Options
Claire Brown Claire Brown
Reply | Threaded
Open this post in threaded view
♦
♦
|

Microscope Air Tables

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I’m being asked by a third party contractor about our microscope air tables.

Do people have their air tables tested and or calibrated once per year?

Sincerely,

Claire
Julian Smith III Julian Smith III
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Microscope Air Tables

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

No--we don't.  If it's still suspended, it's o.k.  We do adjust the
levelers as needed.
Julian

On 2/18/19 9:21 AM, Claire Brown, Dr. wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> I’m being asked by a third party contractor about our microscope air tables.
>
> Do people have their air tables tested and or calibrated once per year?
>
> Sincerely,
>
> Claire


--
Julian P.S. Smith III
Director, Winthrop Microscopy Facility
Dept. of Biology
Winthrop University
349 Columbia Ave
Rock Hill, SCÊ 29733

803-323-2111 x6427 (vox)
803-323-3448 (fax)
803-524-2347 (cell)
Research Website www.birdnest.org/smithj
Personal Website www.rociada-east.net
Sylvie Le Guyader Sylvie Le Guyader
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Microscope Air Tables

In reply to this post by Claire Brown
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

We never do. As long as the tables are floating on all legs (tested by pressing slightly on each corner of the table), we are happy.

Med vänlig hälsning / Best regards

Sylvie

@@@@@@@@@@@@@@@@@@@@@@@@
Sylvie Le Guyader, PhD
Live Cell Imaging Facility Manager
Karolinska Institutet- Bionut Dpt
Hälsovägen 7C,
Room 7362 (lab)/7840 (office)
14157 Huddinge, Sweden
mobile: +46 (0) 73 733 5008
LCI website
Follow our microscopy blog!


-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Claire Brown, Dr.
Sent: den 18 februari 2019 15:22
To: [hidden email]
Subject: Microscope Air Tables

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I'm being asked by a third party contractor about our microscope air tables.

Do people have their air tables tested and or calibrated once per year?

Sincerely,

Claire


När du skickar e-post till Karolinska Institutet (KI) innebär detta att KI kommer att behandla dina personuppgifter. Här finns information om hur KI behandlar personuppgifter<https://ki.se/medarbetare/integritetsskyddspolicy>.


Sending email to Karolinska Institutet (KI) will result in KI processing your personal data. You can read more about KI’s processing of personal data here<https://ki.se/en/staff/data-protection-policy>.
Mark A. Sanders Mark A. Sanders
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Microscope Air Tables

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

As has been indicated, checking for normal operation is always a good idea. Another good idea is checking for vibrations not due to the table on an annual basis. Fans, buildings, cabling all evolve and can induce vibration that can impact the imaging in spite of a functioning table.

Cheers,
Mark

--
Mark Sanders          University of Minnesota
Program Director          Twin Cities Campus
University Imaging Centers
Nikon Center of Excellence
uic.umn.edu
uic.umn.edu/content/locations
Any work utilizing UIC equipment or staff support must acknowledge the University Imaging Centers in publications and presentations.

> On Feb 18, 2019, at 10:15 AM, Sylvie Le Guyader <[hidden email]> wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> We never do. As long as the tables are floating on all legs (tested by pressing slightly on each corner of the table), we are happy.
>
> Med vänlig hälsning / Best regards
>
> Sylvie
>
> @@@@@@@@@@@@@@@@@@@@@@@@
> Sylvie Le Guyader, PhD
> Live Cell Imaging Facility Manager
> Karolinska Institutet- Bionut Dpt
> Hälsovägen 7C,
> Room 7362 (lab)/7840 (office)
> 14157 Huddinge, Sweden
> mobile: +46 (0) 73 733 5008
> LCI website
> Follow our microscopy blog!
>
>
> -----Original Message-----
> From: Confocal Microscopy List <[hidden email]> On Behalf Of Claire Brown, Dr.
> Sent: den 18 februari 2019 15:22
> To: [hidden email]
> Subject: Microscope Air Tables
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> I'm being asked by a third party contractor about our microscope air tables.
>
> Do people have their air tables tested and or calibrated once per year?
>
> Sincerely,
>
> Claire
>
>
> När du skickar e-post till Karolinska Institutet (KI) innebär detta att KI kommer att behandla dina personuppgifter. Här finns information om hur KI behandlar personuppgifter<https://ki.se/medarbetare/integritetsskyddspolicy>.
>
>
> Sending email to Karolinska Institutet (KI) will result in KI processing your personal data. You can read more about KI’s processing of personal data here<https://ki.se/en/staff/data-protection-policy>.
Armstrong, Brian Armstrong, Brian
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Microscope Air Tables

In reply to this post by Claire Brown
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Claire, no, we do not.
Cheers,

Brian Armstrong PhD
Associate Research Professor
Director, Light Microscopy Core Facility
City of Hope, Biomedical Research Center
626-256-4673 x 81259 (BRC) x82872 (Miller) x82874 (Furth)


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Claire Brown, Dr.
Sent: Monday, February 18, 2019 6:22 AM
To: [hidden email]
Subject: Microscope Air Tables

[Attention: This email came from an external source. Do not open attachments or click on links from unknown senders or unexpected emails.]





*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I'm being asked by a third party contractor about our microscope air tables.

Do people have their air tables tested and or calibrated once per year?

Sincerely,

Claire

------------------------------------------------------------
-SECURITY/CONFIDENTIALITY WARNING-

This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. (LCP301)
------------------------------------------------------------
Christopher Yip Christopher Yip
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Microscope Air Tables

In reply to this post by Claire Brown
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

for us - same - as long as they are floating….


Christopher Yip PhD, P.Eng,

Associate Vice President - International Partnerships

Professor
Institute of Biomaterials and Biomedical Engineering
Department of Chemical Engineering and Applied Chemistry
Department of Biochemistry
The Terrence Donnelly Centre for Cellular and Biomolecular Research
University of Toronto
404 - 160 College St
Toronto, Ontario, CANADA M5S 3E1
(416) 978-7853 (office)
(416) 978-4317 (fax)
[hidden email]<mailto:[hidden email]>
http://bigten.med.utoronto.ca
http://www.thedonnellycentre.utoronto.ca

On Feb 18, 2019, at 9:21 AM, Claire Brown, Dr. <[hidden email]<mailto:[hidden email]>> wrote:

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I’m being asked by a third party contractor about our microscope air tables.

Do people have their air tables tested and or calibrated once per year?

Sincerely,

Claire

Arvydas Matiukas Arvydas Matiukas
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: (EXTERNAL) Re: Microscope Air Tables

In reply to this post by Mark A. Sanders
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I would add that different level of vibration dampening may be required
depending on imaging application and especially

pixel resolution . I am now building a custom STED system and wonder how
much more dampening will be required. Regarding
 regular confocal and widefield systems I second what other people told
- as long as table is floating and leveled no further tests

are usually needed.

Best,
Arvydas
----------------------------






Arvydas Matiukas, Ph.D.
Director of Advanced Fluorescence Imaging  Core
SUNY Upstate Medical University
Neuroscience & Physiology Dept


>>> FirstName LastName <[hidden email]> 02/18/19 1:55 PM >>>
*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=S1f_gP6JBtQa_n9ROTrbmAHmOHmC1RGBQeoEqA4i3PA&s=Tjk3SgR0hOlpcVr5LuKOpgnJK4Oy9Q2nm1Ajae4DAgY&e=
Post images on
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=S1f_gP6JBtQa_n9ROTrbmAHmOHmC1RGBQeoEqA4i3PA&s=BQhcAMeTCk7GWXMPPpLyKuYFT8Dx0Vq1SJCWRErgJCs&e=
and include the link in your posting.
*****

As has been indicated, checking for normal operation is always a good
idea. Another good idea is checking for vibrations not due to the table
on an annual basis. Fans, buildings, cabling all evolve and can induce
vibration that can impact the imaging in spite of a functioning table.

Cheers,
Mark

--
Mark Sanders          University of Minnesota
Program Director          Twin Cities Campus
University Imaging Centers
Nikon Center of Excellence
uic.umn.edu
uic.umn.edu/content/locations
Any work utilizing UIC equipment or staff support must acknowledge the
University Imaging Centers in publications and presentations.

> On Feb 18, 2019, at 10:15 AM, Sylvie Le Guyader
<[hidden email]> wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=S1f_gP6JBtQa_n9ROTrbmAHmOHmC1RGBQeoEqA4i3PA&s=Tjk3SgR0hOlpcVr5LuKOpgnJK4Oy9Q2nm1Ajae4DAgY&e=
> Post images on
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=S1f_gP6JBtQa_n9ROTrbmAHmOHmC1RGBQeoEqA4i3PA&s=BQhcAMeTCk7GWXMPPpLyKuYFT8Dx0Vq1SJCWRErgJCs&e=
and include the link in your posting.
> *****
>
> We never do. As long as the tables are floating on all legs (tested by
pressing slightly on each corner of the table), we are happy.

>
> Med vänlig hälsning / Best regards
>
> Sylvie
>
> @@@@@@@@@@@@@@@@@@@@@@@@
> Sylvie Le Guyader, PhD
> Live Cell Imaging Facility Manager
> Karolinska Institutet- Bionut Dpt
> Hälsovägen 7C,
> Room 7362 (lab)/7840 (office)
> 14157 Huddinge, Sweden
> mobile: +46 (0) 73 733 5008
> LCI website
> Follow our microscopy blog!
>
>
> -----Original Message-----
> From: Confocal Microscopy List <[hidden email]> On
Behalf Of Claire Brown, Dr.
> Sent: den 18 februari 2019 15:22
> To: [hidden email]
> Subject: Microscope Air Tables
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=S1f_gP6JBtQa_n9ROTrbmAHmOHmC1RGBQeoEqA4i3PA&s=Tjk3SgR0hOlpcVr5LuKOpgnJK4Oy9Q2nm1Ajae4DAgY&e=
> Post images on
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=S1f_gP6JBtQa_n9ROTrbmAHmOHmC1RGBQeoEqA4i3PA&s=BQhcAMeTCk7GWXMPPpLyKuYFT8Dx0Vq1SJCWRErgJCs> I'm being asked by a third party contractor about our microscope air
tables.
>
> Do people have their air tables tested and or calibrated once per
year?
>
> Sincerely,
>
> Claire
>
>
> När du skickar e-post till Karolinska Institutet (KI) innebär detta
att KI kommer att behandla dina personuppgifter. Här finns information
om hur KI behandlar
personuppgifter<https://urldefense.proofpoint.com/v2/url?u=https-3A__ki.se_medarbetare_integritetsskyddspolicy&d=DwIFaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=S1f_gP6JBtQa_n9ROTrbmAHmOHmC1RGBQeoEqA4i3PA&s=L2mFtaIrFNPj0xAtK0oiV-T3v3eHcOV7-dqxKBohW44&e=>.
>
>
> Sending email to Karolinska Institutet (KI) will result in KI
processing your personal data. You can read more about KI’s processing
of personal data
here<https://urldefense.proofpoint.com/v2/url?u=https-3A__ki.se_en_staff_data-2Dprotection-2Dpolicy&d=DwIFaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=S1f_gP6JBtQa_n9ROTrbmAHmOHmC1RGBQeoEqA4i3PA&s=XWy73d_B5h_4RlX7ylDs0_mc-BlvZe1a5u-eqwDBcEs&e=>.
Sylvie Le Guyader Sylvie Le Guyader
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: (EXTERNAL) Re: Microscope Air Tables

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi everyone

We have antivibration tables and have had no problem so far in our old building. We moved one year ago to a more bling bling building with fantastic ventilation. The ventilation vents and pipes that are above us in the ceiling can create some regular background noise at times. The sound waves nicely push on the antivibration tables the way I would do if I wanted to test the tables. Our STORM and our cell picker system vibrate in unison... How cool is that?

Therefore when building a room, I would in retrospect also check if what is hidden under the ceiling can be noisy. :(

Med vänlig hälsning / Best regards

Sylvie

@@@@@@@@@@@@@@@@@@@@@@@@
Sylvie Le Guyader, PhD
Live Cell Imaging Facility Manager
Karolinska Institutet- Bionut Dpt
Hälsovägen 7C,
Room 7362 (lab)/7840 (office)
14157 Huddinge, Sweden
mobile: +46 (0) 73 733 5008
LCI website
Follow our microscopy blog!



-----Original Message-----
From: Confocal Microscopy List <[hidden email]> On Behalf Of Arvydas Matiukas
Sent: den 21 februari 2019 04:19
To: [hidden email]
Subject: Re: (EXTERNAL) Re: Microscope Air Tables

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

I would add that different level of vibration dampening may be required depending on imaging application and especially

pixel resolution . I am now building a custom STED system and wonder how much more dampening will be required. Regarding  regular confocal and widefield systems I second what other people told
- as long as table is floating and leveled no further tests

are usually needed.

Best,
Arvydas
----------------------------






Arvydas Matiukas, Ph.D.
Director of Advanced Fluorescence Imaging  Core SUNY Upstate Medical University Neuroscience & Physiology Dept


>>> FirstName LastName <[hidden email]> 02/18/19 1:55 PM >>>
*****
To join, leave or search the confocal microscopy listserv, go to:
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=S1f_gP6JBtQa_n9ROTrbmAHmOHmC1RGBQeoEqA4i3PA&s=Tjk3SgR0hOlpcVr5LuKOpgnJK4Oy9Q2nm1Ajae4DAgY&e=
Post images on
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=S1f_gP6JBtQa_n9ROTrbmAHmOHmC1RGBQeoEqA4i3PA&s=BQhcAMeTCk7GWXMPPpLyKuYFT8Dx0Vq1SJCWRErgJCs&e=
and include the link in your posting.
*****

As has been indicated, checking for normal operation is always a good idea. Another good idea is checking for vibrations not due to the table on an annual basis. Fans, buildings, cabling all evolve and can induce vibration that can impact the imaging in spite of a functioning table.

Cheers,
Mark

--
Mark Sanders          University of Minnesota
Program Director          Twin Cities Campus
University Imaging Centers
Nikon Center of Excellence
uic.umn.edu
uic.umn.edu/content/locations
Any work utilizing UIC equipment or staff support must acknowledge the University Imaging Centers in publications and presentations.

> On Feb 18, 2019, at 10:15 AM, Sylvie Le Guyader
<[hidden email]> wrote:
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=S1f_gP6JBtQa_n9ROTrbmAHmOHmC1RGBQeoEqA4i3PA&s=Tjk3SgR0hOlpcVr5LuKOpgnJK4Oy9Q2nm1Ajae4DAgY&e=
> Post images on
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=S1f_gP6JBtQa_n9ROTrbmAHmOHmC1RGBQeoEqA4i3PA&s=BQhcAMeTCk7GWXMPPpLyKuYFT8Dx0Vq1SJCWRErgJCs&e=
and include the link in your posting.
> *****
>
> We never do. As long as the tables are floating on all legs (tested by
pressing slightly on each corner of the table), we are happy.

>
> Med vänlig hälsning / Best regards
>
> Sylvie
>
> @@@@@@@@@@@@@@@@@@@@@@@@
> Sylvie Le Guyader, PhD
> Live Cell Imaging Facility Manager
> Karolinska Institutet- Bionut Dpt
> Hälsovägen 7C,
> Room 7362 (lab)/7840 (office)
> 14157 Huddinge, Sweden
> mobile: +46 (0) 73 733 5008
> LCI website
> Follow our microscopy blog!
>
>
> -----Original Message-----
> From: Confocal Microscopy List <[hidden email]> On
Behalf Of Claire Brown, Dr.
> Sent: den 18 februari 2019 15:22
> To: [hidden email]
> Subject: Microscope Air Tables
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
>
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIFaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=S1f_gP6JBtQa_n9ROTrbmAHmOHmC1RGBQeoEqA4i3PA&s=Tjk3SgR0hOlpcVr5LuKOpgnJK4Oy9Q2nm1Ajae4DAgY&e=
> Post images on
https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIFaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=S1f_gP6JBtQa_n9ROTrbmAHmOHmC1RGBQeoEqA4i3PA&s=BQhcAMeTCk7GWXMPPpLyKuYFT8Dx0Vq1SJCWRErgJCs> I'm being asked by a third party contractor about our microscope air tables.
>
> Do people have their air tables tested and or calibrated once per
year?
>
> Sincerely,
>
> Claire
>
>
> När du skickar e-post till Karolinska Institutet (KI) innebär detta
att KI kommer att behandla dina personuppgifter. Här finns information om hur KI behandlar personuppgifter<https://urldefense.proofpoint.com/v2/url?u=https-3A__ki.se_medarbetare_integritetsskyddspolicy&d=DwIFaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=S1f_gP6JBtQa_n9ROTrbmAHmOHmC1RGBQeoEqA4i3PA&s=L2mFtaIrFNPj0xAtK0oiV-T3v3eHcOV7-dqxKBohW44&e=>.
>
>
> Sending email to Karolinska Institutet (KI) will result in KI
processing your personal data. You can read more about KI’s processing of personal data here<https://urldefense.proofpoint.com/v2/url?u=https-3A__ki.se_en_staff_data-2Dprotection-2Dpolicy&d=DwIFaQ&c=ogn2iPkgF7TkVSicOVBfKg&r=DtXPX1Vw9uh8rqlCEC9BTBr2oYBP4dEG1ecsgK6l-4k&m=S1f_gP6JBtQa_n9ROTrbmAHmOHmC1RGBQeoEqA4i3PA&s=XWy73d_B5h_4RlX7ylDs0_mc-BlvZe1a5u-eqwDBcEs&e=>.


När du skickar e-post till Karolinska Institutet (KI) innebär detta att KI kommer att behandla dina personuppgifter. Här finns information om hur KI behandlar personuppgifter<https://ki.se/medarbetare/integritetsskyddspolicy>.


Sending email to Karolinska Institutet (KI) will result in KI processing your personal data. You can read more about KI’s processing of personal data here<https://ki.se/en/staff/data-protection-policy>.
daj1u06 daj1u06
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Microscope Air Tables

In reply to this post by Claire Brown
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

The only caveat to this would be that if it is powered by a compressor, the pressure vessel of the compressor would likely be subject to a regular statutory inspection / testing (as is in workplace) and if powered from a compressed air cylinder, those also have a regular statutory inspection cycle (in the UK diving air cylinders have a 2.5 year cycle of visual internal inspection and a 5 year hydraulic testing). I guess that most cylinders in the air table scenario just sit there for many years without a thought as to their internal condition.