Heping Yuan |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, sorry if this has been discussed before (which I'm sure it has) but I couldn't find any relevant discussions through search. I was wondering how does one go about measuring field of view on a custom designed multiphoton microscope? Thanks, Tim |
|
Marc Reinig |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** If you have a stage that is calibrated, find a target, move it to the edge of your FOV, note the position, move it to the opposite edge of your FOV, note the position, the difference is your linear FOV. Marco Marco Marc R. Reinig W. M. Keck Center for Adaptive Optical Microscopy University of California Santa Cruz On Wed, Nov 30, 2016 at 11:01 AM, Tim <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi all, sorry if this has been discussed before (which I'm sure it has) but I couldn't find any relevant discussions through search. I was wondering how does one go about measuring field of view on a custom designed multiphoton microscope? > > Thanks, > Tim |
|
Gary G. Li-3 |
|
|
In reply to this post by Heping Yuan
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Look at something with a known size, e.g., a fluorescent slide with known grid spacing. On Wed, Nov 30, 2016 at 2:01 PM, Tim <[hidden email]> wrote: ***** |
|
Heping Yuan |
|
|
In reply to this post by Heping Yuan
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thanks Gary. Would such a slide be commercially available? I did not see any slides with fluorescent grids after doing a search. |
|
George McNamara |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Tim, If you are attending ASCB, you can pick up a box with several different color fluorescence slides from the Chroma Technology booth (if not, you could email them and very politley ask if they would mail a box to you ... might help to offer to pay shipping). For a grid - just stick a hemacytometer on the microscope. If you want to get fancy, put a (dilute) solution of fluorescein or a dilute solution (to not waste money) of fluorescent antibody on it. You can focus in brightfield. enjoy, George On 11/30/2016 2:51 PM, Tim wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Thanks Gary. Would such a slide be commercially available? I did not see any slides with fluorescent grids after doing a search. -- George McNamara, PhD Houston, TX 77054 [hidden email] https://www.linkedin.com/in/georgemcnamara https://works.bepress.com/gmcnamara/75/ http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650 |
|
Gary G. Li-2 |
|
|
In reply to this post by Heping Yuan
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
There are some commercial fluorescent targets with known grid patterns, for example, http://www.maxlevy.com/category/fluorescent-tragets.html Please make sure you select one that suits for your purpose. On Wed, Nov 30, 2016 at 3:51 PM, Tim <[hidden email]> wrote: ***** |
|
Hongtao Chen |
|
|
In reply to this post by George McNamara
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
On Wed, Nov 30, 2016 at 1:31 PM, George McNamara <[hidden email]> wrote: ***** -- Best Regards, Hongtao Chen |
|
Reece, Jeff (NIH/NIDDK) [E] |
|
|
In reply to this post by Gary G. Li-3
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Hi Tim, If you are happy with precision of a few percent, you could use
Powerpoint or similar software, and a decent printer, to print out your own crude sample on regular paper, and image it in transmitted or reflected light.
The spacing on my sample goes down to 500 microns, but your printer might do much better. Cheers, Jeff From: Gary
G. Li [mailto:[hidden email]] ***** To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on
http://www.imgur.com and include the link in your posting. *****
Look at something with a known size, e.g., a fluorescent slide with known grid spacing. On Wed, Nov 30, 2016 at 2:01 PM, Tim <[hidden email]> wrote:
|
|
Gary G. Li-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
You can also paint the printed grid with a fluorescent marker, to make it a crude fluorescent target. On Wed, Nov 30, 2016 at 5:52 PM, Reece, Jeff (NIH/NIDDK) [E] <[hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/ |
|
Tamara Howard |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
If you have a spare etched micrometer slide, you could "paint" the etched lines with the fluorescent marker. We have a very old, etched line slide that someone probably modified to be fluorescent at some time. Swipe the marker (or make a dye paint) over the etched area, let it dry, then carefully wipe the surface clean & at least some of the fluor should stay in the etched lines.
Tamara
...................................................
Tamara Howard Dept. of Cell Biology & Physiology University of New Mexico Albuquerque, NM From: Confocal Microscopy List <[hidden email]> on behalf of Gary G. Li <[hidden email]>
Sent: Wednesday, November 30, 2016 4:09 PM To: [hidden email] Subject: Re: Multiphoton microscope - Field of View ***** To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on
http://www.imgur.com and include the link in your posting. *****
You can also paint the printed grid with a fluorescent marker, to make it a crude fluorescent target. On Wed, Nov 30, 2016 at 5:52 PM, Reece, Jeff (NIH/NIDDK) [E]
<[hidden email]> wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/ |
|||||
|
Benjamin Smith |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
We use the following grid micrometer with a layer of fluorescein and a coverslip: http://www.edmundoptics.com/test-targets/image-analysis-test-targets/multi-grid-standard-stage-micrometer/58607/ The fluorescence reflects off of the mirrored grid nicely, and allows for the width and height to be measured simultaneously. It is also really helpful for aligning a brightfield system to an MP scan field, including any transformations. Cheers, Ben Smith On Wed, Nov 30, 2016 at 4:16 PM, Tamara Howard <[hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/ Benjamin E. Smith, Ph. D. Microscopy Specialist, Vision Science University of California, Berkeley 142 Life Sciences Addition Berkeley, CA 94720-3200 Tel (510) 642-9712 Fax (510) 643-6791 e-mail: [hidden email] http://vision.berkeley.edu/ |
|||||
|
Bill Bennett |
|
|
In reply to this post by Heping Yuan
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** We use the Convollaria (lily of the valley) root cross-section slide that came with our old Zeiss LSM500. It gives a beautiful bright-field image, and fluoresces intensely right across most of our Mai Tai wavelengths. So you can use it to align BF and 2P captures and transformations, as previously mentioned. But we also find it very useful as a startup check for our 2P system to check detectors, beam alignment etc. We got ours with the Zeiss, but I believe you can buy these slides quite cheaply e.g. http://www.msmedia.com.au/convallaria-rhizome-training-slide-stained-for-fluorescence.php Bill Bennett |