Hi all, sorry if this has been discussed before (which I'm sure it has) but I couldn't find any relevant discussions through search. I was wondering how does one go about measuring field of view on a custom designed multiphoton microscope?

Thanks,
Tim
Marc Reinig Marc Reinig
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Multiphoton microscope - Field of View

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

If you have a stage that is calibrated, find a target, move it to the
edge of your FOV, note the position, move it to the opposite edge of
your FOV, note the position, the difference is your linear FOV.

Marco
Marco

Marc R. Reinig
W. M. Keck Center for Adaptive Optical Microscopy
University of California Santa Cruz



On Wed, Nov 30, 2016 at 11:01 AM, Tim <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi all, sorry if this has been discussed before (which I'm sure it has) but I couldn't find any relevant discussions through search. I was wondering how does one go about measuring field of view on a custom designed multiphoton microscope?
>
> Thanks,
> Tim
Gary G. Li-3 Gary G. Li-3
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Multiphoton microscope - Field of View

In reply to this post by Heping Yuan
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****


Look at something with a known size, e.g., a fluorescent slide with known grid spacing.  


On Wed, Nov 30, 2016 at 2:01 PM, Tim <[hidden email]> wrote:
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi all, sorry if this has been discussed before (which I'm sure it has) but I couldn't find any relevant discussions through search. I was wondering how does one go about measuring field of view on a custom designed multiphoton microscope?

Thanks,
Tim

Heping Yuan Heping Yuan
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Multiphoton microscope - Field of View

In reply to this post by Heping Yuan
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Thanks Gary. Would such a slide be commercially available? I did not see any slides with fluorescent grids after doing a search.
George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Multiphoton microscope - Field of View

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Tim,

If you are attending ASCB, you can pick up a box with several different
color fluorescence slides from the Chroma Technology booth (if not, you
could email them and very politley ask if they would mail a box to you
... might help to offer to pay shipping).

For a grid - just stick a hemacytometer on the microscope. If you want
to get fancy, put a (dilute) solution of fluorescein or a dilute
solution (to not waste money) of fluorescent antibody on it. You can
focus in brightfield.

enjoy,

George


On 11/30/2016 2:51 PM, Tim wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Thanks Gary. Would such a slide be commercially available? I did not see any slides with fluorescent grids after doing a search.

--


George McNamara, PhD
Houston, TX 77054
[hidden email]
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/
http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650
Gary G. Li-2 Gary G. Li-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Multiphoton microscope - Field of View

In reply to this post by Heping Yuan
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

There are some commercial fluorescent targets with known grid patterns, for example,  


http://www.maxlevy.com/category/fluorescent-tragets.html


Please make sure you select one that suits for your purpose.



On Wed, Nov 30, 2016 at 3:51 PM, Tim <[hidden email]> wrote:
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Thanks Gary. Would such a slide be commercially available? I did not see any slides with fluorescent grids after doing a search.

Hongtao Chen Hongtao Chen
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Multiphoton microscope - Field of View

In reply to this post by George McNamara
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

On Wed, Nov 30, 2016 at 1:31 PM, George McNamara <[hidden email]> wrote:
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Tim,

If you are attending ASCB, you can pick up a box with several different color fluorescence slides from the Chroma Technology booth (if not, you could email them and very politley ask if they would mail a box to you ... might help to offer to pay shipping).

For a grid - just stick a hemacytometer on the microscope. If you want to get fancy, put a (dilute) solution of fluorescein or a dilute solution (to not waste money) of fluorescent antibody on it. You can focus in brightfield.

enjoy,

George



On 11/30/2016 2:51 PM, Tim wrote:
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Thanks Gary. Would such a slide be commercially available? I did not see any slides with fluorescent grids after doing a search.

--


George McNamara, PhD
Houston, TX 77054
[hidden email]
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/
http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650



--
Best Regards,
Hongtao Chen
Reece, Jeff (NIH/NIDDK) [E] Reece, Jeff (NIH/NIDDK) [E]
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Multiphoton microscope - Field of View

In reply to this post by Gary G. Li-3
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Tim,

 

If you are happy with precision of a few percent, you could use Powerpoint or similar software, and a decent printer, to print out your own crude sample on regular paper, and image it in transmitted or reflected light.  The spacing on my sample goes down to 500 microns, but your printer might do much better.

 

Cheers,

Jeff

 

From: Gary G. Li [mailto:[hidden email]]
Sent: Wednesday, November 30, 2016 3:00 PM
To: [hidden email]
Subject: Re: Multiphoton microscope - Field of View

 

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

 

Look at something with a known size, e.g., a fluorescent slide with known grid spacing.  

 

On Wed, Nov 30, 2016 at 2:01 PM, Tim <[hidden email]> wrote:

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi all, sorry if this has been discussed before (which I'm sure it has) but I couldn't find any relevant discussions through search. I was wondering how does one go about measuring field of view on a custom designed multiphoton microscope?

Thanks,
Tim

 

Gary G. Li-2 Gary G. Li-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Multiphoton microscope - Field of View

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****


You can also paint the printed grid with a fluorescent marker, to make it a crude fluorescent target.   



On Wed, Nov 30, 2016 at 5:52 PM, Reece, Jeff (NIH/NIDDK) [E] <[hidden email]> wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Tim,

 

If you are happy with precision of a few percent, you could use Powerpoint or similar software, and a decent printer, to print out your own crude sample on regular paper, and image it in transmitted or reflected light.  The spacing on my sample goes down to 500 microns, but your printer might do much better.

 

Cheers,

Jeff

 

From: Gary G. Li [mailto:[hidden email]]
Sent: Wednesday, November 30, 2016 3:00 PM
To: [hidden email]
Subject: Re: Multiphoton microscope - Field of View

 

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

 

Look at something with a known size, e.g., a fluorescent slide with known grid spacing.  

 

On Wed, Nov 30, 2016 at 2:01 PM, Tim <[hidden email]> wrote:

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi all, sorry if this has been discussed before (which I'm sure it has) but I couldn't find any relevant discussions through search. I was wondering how does one go about measuring field of view on a custom designed multiphoton microscope?

Thanks,
Tim

 


Tamara Howard Tamara Howard
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Multiphoton microscope - Field of View

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

If you have a spare etched micrometer slide, you could "paint" the etched lines with the fluorescent marker. We have a very old, etched line slide that someone probably modified to be fluorescent at some time. Swipe the marker (or make a dye paint) over the etched area, let it dry, then carefully wipe the surface clean & at least some of the fluor should stay in the etched lines.


Tamara


...................................................
Tamara Howard
Dept. of Cell Biology & Physiology
University of New Mexico
Albuquerque, NM




From: Confocal Microscopy List <[hidden email]> on behalf of Gary G. Li <[hidden email]>
Sent: Wednesday, November 30, 2016 4:09 PM
To: [hidden email]
Subject: Re: Multiphoton microscope - Field of View
 
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
lists.umn.edu
[hidden email]: listserv archives. confocalmicroscopy
www.imgur.com
Imgur is the best place to share and enjoy the most awesome images on the Internet. Every day, millions of people use Imgur to be entertained and inspired by funny ...


You can also paint the printed grid with a fluorescent marker, to make it a crude fluorescent target.   



On Wed, Nov 30, 2016 at 5:52 PM, Reece, Jeff (NIH/NIDDK) [E] <[hidden email]> wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Tim,

 

If you are happy with precision of a few percent, you could use Powerpoint or similar software, and a decent printer, to print out your own crude sample on regular paper, and image it in transmitted or reflected light.  The spacing on my sample goes down to 500 microns, but your printer might do much better.

 

Cheers,

Jeff

 

From: Gary G. Li [mailto:[hidden email]]
Sent: Wednesday, November 30, 2016 3:00 PM
To: [hidden email]
Subject: Re: Multiphoton microscope - Field of View

 

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

 

Look at something with a known size, e.g., a fluorescent slide with known grid spacing.  

 

On Wed, Nov 30, 2016 at 2:01 PM, Tim <[hidden email]> wrote:

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi all, sorry if this has been discussed before (which I'm sure it has) but I couldn't find any relevant discussions through search. I was wondering how does one go about measuring field of view on a custom designed multiphoton microscope?

Thanks,
Tim

 


Benjamin Smith Benjamin Smith
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Multiphoton microscope - Field of View

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
We use the following grid micrometer with a layer of fluorescein and a coverslip: http://www.edmundoptics.com/test-targets/image-analysis-test-targets/multi-grid-standard-stage-micrometer/58607/

The fluorescence reflects off of the mirrored grid nicely, and allows for the width and height to be measured simultaneously.  It is also really helpful for aligning a brightfield system to an MP scan field, including any transformations.

Cheers,
   Ben Smith

On Wed, Nov 30, 2016 at 4:16 PM, Tamara Howard <[hidden email]> wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

If you have a spare etched micrometer slide, you could "paint" the etched lines with the fluorescent marker. We have a very old, etched line slide that someone probably modified to be fluorescent at some time. Swipe the marker (or make a dye paint) over the etched area, let it dry, then carefully wipe the surface clean & at least some of the fluor should stay in the etched lines.


Tamara


...................................................
Tamara Howard
Dept. of Cell Biology & Physiology
University of New Mexico
Albuquerque, NM




From: Confocal Microscopy List <[hidden email]> on behalf of Gary G. Li <[hidden email]>
Sent: Wednesday, November 30, 2016 4:09 PM
To: [hidden email]
Subject: Re: Multiphoton microscope - Field of View
 
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Imgur is the best place to share and enjoy the most awesome images on the Internet. Every day, millions of people use Imgur to be entertained and inspired by funny ...


You can also paint the printed grid with a fluorescent marker, to make it a crude fluorescent target.   



On Wed, Nov 30, 2016 at 5:52 PM, Reece, Jeff (NIH/NIDDK) [E] <[hidden email]> wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Tim,

 

If you are happy with precision of a few percent, you could use Powerpoint or similar software, and a decent printer, to print out your own crude sample on regular paper, and image it in transmitted or reflected light.  The spacing on my sample goes down to 500 microns, but your printer might do much better.

 

Cheers,

Jeff

 

From: Gary G. Li [mailto:[hidden email]]
Sent: Wednesday, November 30, 2016 3:00 PM
To: [hidden email]
Subject: Re: Multiphoton microscope - Field of View

 

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

 

Look at something with a known size, e.g., a fluorescent slide with known grid spacing.  

 

On Wed, Nov 30, 2016 at 2:01 PM, Tim <[hidden email]> wrote:

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi all, sorry if this has been discussed before (which I'm sure it has) but I couldn't find any relevant discussions through search. I was wondering how does one go about measuring field of view on a custom designed multiphoton microscope?

Thanks,
Tim

 





--
Benjamin E. Smith, Ph. D.
Microscopy Specialist, Vision Science
University of California, Berkeley
142 Life Sciences Addition
Berkeley, CA  94720-3200
Tel  (510) 642-9712
Fax (510) 643-6791
e-mail: [hidden email]
http://vision.berkeley.edu/
Bill Bennett Bill Bennett
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Multiphoton microscope - Field of View

In reply to this post by Heping Yuan
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

We use the Convollaria (lily of the valley) root cross-section slide that came with our old Zeiss LSM500.   It gives a beautiful bright-field image, and fluoresces intensely right across most of our Mai Tai wavelengths.   So you can use it to align BF and 2P captures and transformations, as previously mentioned.  But we also find it very useful as a startup check for our 2P system to check detectors,  beam alignment etc.

We got ours with the Zeiss, but I believe you can buy these slides quite cheaply e.g.
http://www.msmedia.com.au/convallaria-rhizome-training-slide-stained-for-fluorescence.php

Bill Bennett