To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Microscopists, for our multiphoton scope we use an acousto-optical modulator for intensity modulation. Those AOM have a nice overall throughput (>85%) and a high bandwidth (>1MHz), however they add a lot of group delay dispersion (GDD) to our setup - typically they use a 1cm - 2cm TeO2 crystal for the NIR wavelengths from 700nm to 1000nm. For our multiphoton setup, the AOM is the main contributor of GDD right now. I was wondering: what is your preferred choice for intensity modulation and what are the respective advantages/disadvantages? Thanks,Peter |
|
Benjamin Smith |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** For linear scanning 2P microscopes, an achromatic half-wave plate (in a motorized rotating mount) paired with a glan-laser polarizer is a clear winner. Our measured contrast ratio with this attentuator setup is almost 10,000:1, which is pretty impressive when you consider that the spec sheet for the laser states a polarization ratio of only >500:1 - guess we got a good one. Additionally, since the half wave plate is achromatic, the attenuation is more or less flat across the entire tuning spectrum of the laser. The reason this setup works specifically for linear scanning microscopes is that the angular velocity of a linear scanning mirror is very constant. Therefore, the laser intensity will be uniform across the scan field, negating the need for flyback blanking unless it is absolutely critical not to have the laser shine outside of your scan field. On our systems, the flyback over-scan is about 50% the width of the imaging scan field (25% on each side). In comparison, a Pockel's cell on a linear scanning system adds more cost, more operational complexity, and less performance, for no real gain. That said, if you have a resonant scanner, than your only real choice is a Pockel's cell. This is because since a resonant mirror is under constant angular acceleration, the mirror stalls out at the ends of the scan for an appreciable amount of time. The end result being that if the 2P laser was left on for the full resonant sweep, you would see that the laser is much brighter at the edges of the scan field due to the increased dwell time. Therefore a Pockel's cell is needed to rapidly attenuate the laser at the end of each sweep of the mirror to keep the laser from burning the tissue. One thing that surprised me, the GVD for the KDP crystals in a Pockel's cell is a measly 5.28 fs2/mm at 950 nm which is very low, so even though the path length of a Pockel's cell looks long, the GVD is still very low compared to most optical glasses. You also inspired me to look up the GVD of TeO2 at 950 nm, which is 388 fs2/mm, that's higher than most flint glasses! By my math, a 140 fs pulse going through 2 cm of the stuff gets broadened to 208 fs, and a 70 fs pulse gets broadened to 315 fs. While bad, keep in mind that this is still comparable to the measured GVD in objectives (and likely the scan lens and tube lens, but couldn't find any papers or specs on those): http://users.ox.ac.uk/~atdgroup/publications/Tullis,%20I.D.C.,%20Proc.%20SPIE,%206089,%202006.pdf In summary: Achromatic half wave plate in motorized mount vs. Pockel's cell Pros: Relatively cheap Isn't damaged if left on for long periods of time Spectrally flat attenuation = Very broad spectral bandwidth High contrast ratio Easy to calibrate and use Cons: No fly-back blanking - Cannot be used for resonant 2P scanning microscopes without risk of burning sample outside of scan field One final note, if you do get a Pockel's cell, definitely get a model with an integrated glan-laser polarizer and beam dump. This makes them much simpler to install and safer to align. Cheers, Ben Smith On Mon, Sep 23, 2019 at 11:28 PM Peter Müller < [hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear Microscopists, > for our multiphoton scope we use an acousto-optical modulator for > intensity modulation. Those AOM have a nice overall throughput (>85%) and a > high bandwidth (>1MHz), however they add a lot of group delay dispersion > (GDD) to our setup - typically they use a 1cm - 2cm TeO2 crystal for the > NIR wavelengths from 700nm to 1000nm. For our multiphoton setup, the AOM is > the main contributor of GDD right now. I was wondering: what is your > preferred choice for intensity modulation and what are the respective > advantages/disadvantages? > > Thanks,Peter > -- Benjamin E. Smith, Ph. D. Imaging Specialist, Vision Science University of California, Berkeley 195 Life Sciences Addition Berkeley, CA 94720-3200 Tel (510) 642-9712 Fax (510) 643-6791 e-mail: [hidden email] https://vision.berkeley.edu/faculty/core-grants-nei/core-grant-microscopic-imaging/ |
|
|
In reply to this post by P. Muller
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Peter, I concur with Ben Smith's beautifully detailed reply. I think a half waveplate and polarizer combo is the best and the simplest option for intensity control, and we use it regularly. However, even if you have the half waveplate mounted on a stepper motor, it is not going to be a fast modulation option. I have other worries about AOM's, in addition to the GVD. Since they are diffractive elements, they would introduce spatial chirp if used with a femtosecond beam with a large spectral content ( if we are not using the zero order). However, I have not properly tested them to see if this really matters. May be the spectral chirp is properly compensated by the AOM manufacturers. Has anyone in this group tested it for the spatial chirp? Sudipta On Tue, 24 Sep 2019 06:03:38 +0000, Peter Müller wrote > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear Microscopists, > for our multiphoton scope we use an acousto-optical modulator for > intensity modulation. Those AOM have a nice overall throughput (>85%) > and a high bandwidth (>1MHz), however they add a lot of group delay > dispersion (GDD) to our setup - typically they use a 1cm - 2cm TeO2 > crystal for the NIR wavelengths from 700nm to 1000nm. For our > multiphoton setup, the AOM is the main contributor of GDD right now. > I was wondering: what is your preferred choice for intensity > modulation and what are the respective advantages/disadvantages? > > Thanks,Peter Prof. Sudipta Maiti Dept. of Chemical Sciences Tata Institute of Fundamental Research Homi Bhabha Road, Colaba Mumbai 400005, India Ph. +91 222 278 2716 Alternate e-mail: [hidden email] webpage: biophotonics.co.in |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Sudipta, Ben, thanks for your really nice answers. We do want fast modulation (>1MHz). In fact we have seen some spatial chirp from the AOM, mostly in form of an slightly elliptical beam profile. The blue wavelength components of the laser get less diffracted than the red components and will spatially separate, as you mention. This gets worse with an increased AOM carrier frequency (typ. 80MHz). But if go through the math, you'll notice that you need more than 5 meters of free space propagation for a 100 fs beam at 800nm to significantly separate the beam's components. Best, Peter Am Dienstag, 24. September 2019, 19:57:18 MESZ hat maiti <[hidden email]> Folgendes geschrieben: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Peter, I concur with Ben Smith's beautifully detailed reply. I think a half waveplate and polarizer combo is the best and the simplest option for intensity control, and we use it regularly. However, even if you have the half waveplate mounted on a stepper motor, it is not going to be a fast modulation option. I have other worries about AOM's, in addition to the GVD. Since they are diffractive elements, they would introduce spatial chirp if used with a femtosecond beam with a large spectral content ( if we are not using the zero order). However, I have not properly tested them to see if this really matters. May be the spectral chirp is properly compensated by the AOM manufacturers. Has anyone in this group tested it for the spatial chirp? Sudipta On Tue, 24 Sep 2019 06:03:38 +0000, Peter Müller wrote > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear Microscopists, > for our multiphoton scope we use an acousto-optical modulator for > intensity modulation. Those AOM have a nice overall throughput (>85%) > and a high bandwidth (>1MHz), however they add a lot of group delay > dispersion (GDD) to our setup - typically they use a 1cm - 2cm TeO2 > crystal for the NIR wavelengths from 700nm to 1000nm. For our > multiphoton setup, the AOM is the main contributor of GDD right now. > I was wondering: what is your preferred choice for intensity > modulation and what are the respective advantages/disadvantages? > > Thanks,Peter Prof. Sudipta Maiti Dept. of Chemical Sciences Tata Institute of Fundamental Research Homi Bhabha Road, Colaba Mumbai 400005, India Ph. +91 222 278 2716 Alternate e-mail: [hidden email] webpage: biophotonics.co.in |
|
Mark Cannell-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi All When we used an AOM for 2P intensity modulation we used the zero order beam and with 95% diffraction efficiency the 2P excitation rate was so low when the AOM was fully energized that the zero order beam was effectively shut off. For safety there was also a 2ms response time mechanical shutter that would block the beam when not scanning. That said, we did not expect the AOM to provide full beam intensity control, but had a separate attenuator (reflective ND filter) to set the maximum power desired. HTH Mark Mark Cannell Ph.D. FRSNZ, FISHR University of Bristol, UK On 9/26/19, 7:12 AM, "Confocal Microscopy List on behalf of Peter Müller" <[hidden email] on behalf of [hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Sudipta, Ben, thanks for your really nice answers. We do want fast modulation (>1MHz). In fact we have seen some spatial chirp from the AOM, mostly in form of an slightly elliptical beam profile. The blue wavelength components of the laser get less diffracted than the red components and will spatially separate, as you mention. This gets worse with an increased AOM carrier frequency (typ. 80MHz). But if go through the math, you'll notice that you need more than 5 meters of free space propagation for a 100 fs beam at 800nm to significantly separate the beam's components. Best, Peter Am Dienstag, 24. September 2019, 19:57:18 MESZ hat maiti <[hidden email]> Folgendes geschrieben: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Peter, I concur with Ben Smith's beautifully detailed reply. I think a half waveplate and polarizer combo is the best and the simplest option for intensity control, and we use it regularly. However, even if you have the half waveplate mounted on a stepper motor, it is not going to be a fast modulation option. I have other worries about AOM's, in addition to the GVD. Since they are diffractive elements, they would introduce spatial chirp if used with a femtosecond beam with a large spectral content ( if we are not using the zero order). However, I have not properly tested them to see if this really matters. May be the spectral chirp is properly compensated by the AOM manufacturers. Has anyone in this group tested it for the spatial chirp? Sudipta On Tue, 24 Sep 2019 06:03:38 +0000, Peter Müller wrote > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear Microscopists, > for our multiphoton scope we use an acousto-optical modulator for > intensity modulation. Those AOM have a nice overall throughput (>85%) > and a high bandwidth (>1MHz), however they add a lot of group delay > dispersion (GDD) to our setup - typically they use a 1cm - 2cm TeO2 > crystal for the NIR wavelengths from 700nm to 1000nm. For our > multiphoton setup, the AOM is the main contributor of GDD right now. > I was wondering: what is your preferred choice for intensity > modulation and what are the respective advantages/disadvantages? > > Thanks,Peter Prof. Sudipta Maiti Dept. of Chemical Sciences Tata Institute of Fundamental Research Homi Bhabha Road, Colaba Mumbai 400005, India Ph. +91 222 278 2716 Alternate e-mail: [hidden email] webpage: biophotonics.co.in |
|
Zdenek Svindrych-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Mark, brilliant idea! Now just shoot the beam through the crystal three times and get 8000:1 dynamic range :-). I think I got inspired by some of Stefan Hell's papers here... But you'll need a bit more GDD pre-compensation. On the other hand, with the first order beam (let's call it "the old way") different spectral components will diffract under slightly different angles (but I would not call this a spatially variable chirp, rather very poor temporal focusing). So if all spectral components recombine in the focus, that is, the volume of the crystal is roughly conjugate to the focal plane, there should be no problem. This effect is only slight (depending on your laser bandwidth), but you can still observe it on some microscopes: when you unscrew the lens, the laser beam should be centered in the BFP, when you switch to a different wavelength the beam drifts away as the laser is being tuned, and then jumps back as the crystal driving frequency is adjusted to a new value (this is what I've seen with LSM-780 and Chameleon Vision, but now when I'm thinking about it, it might be confounded by some effects of the internal GDD compensation and beam stabilization elements inside the laser). So even if the "grating effect" of the crystal is small, it should be possible to eliminate it by proper optical design. Best, zdenek On Thu, Sep 26, 2019 at 5:40 AM Mark Cannell <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi All > > When we used an AOM for 2P intensity modulation we used the zero order > beam and with 95% diffraction efficiency the 2P excitation rate was so low > when the AOM was fully energized that the zero order beam was effectively > shut off. For safety there was also a 2ms response time mechanical shutter > that would block the beam when not scanning. That said, we did not expect > the AOM to provide full beam intensity control, but had a separate > attenuator (reflective ND filter) to set the maximum power desired. > > HTH Mark > > Mark Cannell Ph.D. FRSNZ, FISHR > University of Bristol, UK > > On 9/26/19, 7:12 AM, "Confocal Microscopy List on behalf of Peter Müller" > <[hidden email] on behalf of > [hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your > posting. > ***** > > Sudipta, Ben, > > thanks for your really nice answers. > > We do want fast modulation (>1MHz). > > In fact we have seen some spatial chirp from the AOM, mostly in form > of an slightly elliptical beam profile. The blue wavelength components of > the laser get less diffracted than the red components and will spatially > separate, as you mention. This gets worse with an increased AOM carrier > frequency (typ. 80MHz). But if go through the math, you'll notice that you > need more than 5 meters of free space propagation for a 100 fs beam at > 800nm to significantly separate the beam's components. > > Best, > Peter > Am Dienstag, 24. September 2019, 19:57:18 MESZ hat maiti < > [hidden email]> Folgendes geschrieben: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your > posting. > ***** > > Dear Peter, > > I concur with Ben Smith's beautifully detailed reply. I think a half > waveplate and > polarizer combo is the best and the simplest option for intensity > control, and we > use it regularly. However, even if you have the half waveplate mounted > on a > stepper motor, it is not going to be a fast modulation option. > > I have other worries about AOM's, in addition to the GVD. Since they > are > diffractive elements, they would introduce spatial chirp if used with > a > femtosecond beam with a large spectral content ( if we are not using > the zero > order). However, I have not properly tested them to see if this > really matters. > May be the spectral chirp is properly compensated by the AOM > manufacturers. Has > anyone in this group tested it for the spatial chirp? > > Sudipta > > > On Tue, 24 Sep 2019 06:03:38 +0000, Peter Müller wrote > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Dear Microscopists, > > for our multiphoton scope we use an acousto-optical modulator for > > intensity modulation. Those AOM have a nice overall throughput (>85%) > > and a high bandwidth (>1MHz), however they add a lot of group delay > > dispersion (GDD) to our setup - typically they use a 1cm - 2cm TeO2 > > crystal for the NIR wavelengths from 700nm to 1000nm. For our > > multiphoton setup, the AOM is the main contributor of GDD right now. > > I was wondering: what is your preferred choice for intensity > > modulation and what are the respective advantages/disadvantages? > > > > Thanks,Peter > > > Prof. Sudipta Maiti > Dept. of Chemical Sciences > Tata Institute of Fundamental Research > Homi Bhabha Road, Colaba > Mumbai 400005, India > Ph. +91 222 278 2716 > Alternate e-mail: [hidden email] > webpage: biophotonics.co.in > > > > -- -- Zdenek Svindrych, Ph.D. Research Associate - Imaging Specialist Department of Biochemistry and Cell Biology Geisel School of Medicine at Dartmouth |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** GDD compensators can be designed with a retro-reflective trip through all the glass, but this increases the complexity (and cost) of the design. I've done this with a manual design that worked fairly well, but still needed careful initial alignment. Once it was aligned though, tuning the wavelength and adjusting the main prism wouldn't walk the beam. For power control I prefer a Glan Laser splitter and an achromatic zero-order 1/2 waveplate. It gives good extinction and the Glan Laser is air-spaced so it is very tough. When I needed fast modulation I used an EOM combined with the Glan Laser-waveplate arrangement. I used the GL to set the modulation range and let the EOM work within the modulation depth it was capable of. Craig On Thu, Sep 26, 2019 at 1:52 PM Zdenek Svindrych <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Mark, > brilliant idea! Now just shoot the beam through the crystal three times and > get 8000:1 dynamic range :-). I think I got inspired by some of Stefan > Hell's papers here... But you'll need a bit more GDD pre-compensation. > > On the other hand, with the first order beam (let's call it "the old way") > different spectral components will diffract under slightly different angles > (but I would not call this a spatially variable chirp, rather very poor > temporal focusing). So if all spectral components recombine in the focus, > that is, the volume of the crystal is roughly conjugate to the focal plane, > there should be no problem. This effect is only slight (depending on your > laser bandwidth), but you can still observe it on some microscopes: when > you unscrew the lens, the laser beam should be centered in the BFP, when > you switch to a different wavelength the beam drifts away as the laser is > being tuned, and then jumps back as the crystal driving frequency is > adjusted to a new value (this is what I've seen with LSM-780 and Chameleon > Vision, but now when I'm thinking about it, it might be confounded by some > effects of the internal GDD compensation and beam stabilization elements > inside the laser). > > So even if the "grating effect" of the crystal is small, it should be > possible to eliminate it by proper optical design. > > Best, zdenek > > On Thu, Sep 26, 2019 at 5:40 AM Mark Cannell <[hidden email]> > wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Hi All > > > > When we used an AOM for 2P intensity modulation we used the zero order > > beam and with 95% diffraction efficiency the 2P excitation rate was so > low > > when the AOM was fully energized that the zero order beam was effectively > > shut off. For safety there was also a 2ms response time mechanical > shutter > > that would block the beam when not scanning. That said, we did not > expect > > the AOM to provide full beam intensity control, but had a separate > > attenuator (reflective ND filter) to set the maximum power desired. > > > > HTH Mark > > > > Mark Cannell Ph.D. FRSNZ, FISHR > > University of Bristol, UK > > > > On 9/26/19, 7:12 AM, "Confocal Microscopy List on behalf of Peter > Müller" > > <[hidden email] on behalf of > > [hidden email]> wrote: > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > > posting. > > ***** > > > > Sudipta, Ben, > > > > thanks for your really nice answers. > > > > We do want fast modulation (>1MHz). > > > > In fact we have seen some spatial chirp from the AOM, mostly in form > > of an slightly elliptical beam profile. The blue wavelength components of > > the laser get less diffracted than the red components and will spatially > > separate, as you mention. This gets worse with an increased AOM carrier > > frequency (typ. 80MHz). But if go through the math, you'll notice that > you > > need more than 5 meters of free space propagation for a 100 fs beam at > > 800nm to significantly separate the beam's components. > > > > Best, > > Peter > > Am Dienstag, 24. September 2019, 19:57:18 MESZ hat maiti < > > [hidden email]> Folgendes geschrieben: > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > > posting. > > ***** > > > > Dear Peter, > > > > I concur with Ben Smith's beautifully detailed reply. I think a half > > waveplate and > > polarizer combo is the best and the simplest option for intensity > > control, and we > > use it regularly. However, even if you have the half waveplate > mounted > > on a > > stepper motor, it is not going to be a fast modulation option. > > > > I have other worries about AOM's, in addition to the GVD. Since they > > are > > diffractive elements, they would introduce spatial chirp if used with > > a > > femtosecond beam with a large spectral content ( if we are not using > > the zero > > order). However, I have not properly tested them to see if this > > really matters. > > May be the spectral chirp is properly compensated by the AOM > > manufacturers. Has > > anyone in this group tested it for the spatial chirp? > > > > Sudipta > > > > > > On Tue, 24 Sep 2019 06:03:38 +0000, Peter Müller wrote > > > ***** > > > To join, leave or search the confocal microscopy listserv, go to: > > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > > Post images on http://www.imgur.com and include the link in your > > posting. > > > ***** > > > > > > Dear Microscopists, > > > for our multiphoton scope we use an acousto-optical modulator for > > > intensity modulation. Those AOM have a nice overall throughput > (>85%) > > > and a high bandwidth (>1MHz), however they add a lot of group > delay > > > dispersion (GDD) to our setup - typically they use a 1cm - 2cm TeO2 > > > crystal for the NIR wavelengths from 700nm to 1000nm. For our > > > multiphoton setup, the AOM is the main contributor of GDD right > now. > > > I was wondering: what is your preferred choice for intensity > > > modulation and what are the respective advantages/disadvantages? > > > > > > Thanks,Peter > > > > > > Prof. Sudipta Maiti > > Dept. of Chemical Sciences > > Tata Institute of Fundamental Research > > Homi Bhabha Road, Colaba > > Mumbai 400005, India > > Ph. +91 222 278 2716 > > Alternate e-mail: [hidden email] > > webpage: biophotonics.co.in > > > > > > > > > > -- > -- > Zdenek Svindrych, Ph.D. > Research Associate - Imaging Specialist > Department of Biochemistry and Cell Biology > Geisel School of Medicine at Dartmouth > |
|
Mark Cannell-2 |
|
|
In reply to this post by Zdenek Svindrych-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Zedenek Another advantage of using the zero order is that the beam doesn't walk when tuning and we found that a 95% beam attenuation was sufficient to effectively eliminate sample damage during the over scan and fly back (400 fold reduction in excitation rate). I think that some heating effects that can appear in the AOM 1st order bean were also reduced. We also had a pre chirp system in that only needed to be adjusted once for all the optical components so alignment was pretty stable. See https://doi.org/10.1007/s004240050169 Cheers Mark On 9/26/19, 8:52 PM, "Confocal Microscopy List on behalf of Zdenek Svindrych" <[hidden email] on behalf of [hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Mark, brilliant idea! Now just shoot the beam through the crystal three times and get 8000:1 dynamic range :-). I think I got inspired by some of Stefan Hell's papers here... But you'll need a bit more GDD pre-compensation. On the other hand, with the first order beam (let's call it "the old way") different spectral components will diffract under slightly different angles (but I would not call this a spatially variable chirp, rather very poor temporal focusing). So if all spectral components recombine in the focus, that is, the volume of the crystal is roughly conjugate to the focal plane, there should be no problem. This effect is only slight (depending on your laser bandwidth), but you can still observe it on some microscopes: when you unscrew the lens, the laser beam should be centered in the BFP, when you switch to a different wavelength the beam drifts away as the laser is being tuned, and then jumps back as the crystal driving frequency is adjusted to a new value (this is what I've seen with LSM-780 and Chameleon Vision, but now when I'm thinking about it, it might be confounded by some effects of the internal GDD compensation and beam stabilization elements inside the laser). So even if the "grating effect" of the crystal is small, it should be possible to eliminate it by proper optical design. Best, zdenek On Thu, Sep 26, 2019 at 5:40 AM Mark Cannell <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi All > > When we used an AOM for 2P intensity modulation we used the zero order > beam and with 95% diffraction efficiency the 2P excitation rate was so low > when the AOM was fully energized that the zero order beam was effectively > shut off. For safety there was also a 2ms response time mechanical shutter > that would block the beam when not scanning. That said, we did not expect > the AOM to provide full beam intensity control, but had a separate > attenuator (reflective ND filter) to set the maximum power desired. > > HTH Mark > > Mark Cannell Ph.D. FRSNZ, FISHR > University of Bristol, UK > > On 9/26/19, 7:12 AM, "Confocal Microscopy List on behalf of Peter Müller" > <[hidden email] on behalf of > [hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your > posting. > ***** > > Sudipta, Ben, > > thanks for your really nice answers. > > We do want fast modulation (>1MHz). > > In fact we have seen some spatial chirp from the AOM, mostly in form > of an slightly elliptical beam profile. The blue wavelength components of > the laser get less diffracted than the red components and will spatially > separate, as you mention. This gets worse with an increased AOM carrier > frequency (typ. 80MHz). But if go through the math, you'll notice that you > need more than 5 meters of free space propagation for a 100 fs beam at > 800nm to significantly separate the beam's components. > > Best, > Peter > Am Dienstag, 24. September 2019, 19:57:18 MESZ hat maiti < > [hidden email]> Folgendes geschrieben: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your > posting. > ***** > > Dear Peter, > > I concur with Ben Smith's beautifully detailed reply. I think a half > waveplate and > polarizer combo is the best and the simplest option for intensity > control, and we > use it regularly. However, even if you have the half waveplate mounted > on a > stepper motor, it is not going to be a fast modulation option. > > I have other worries about AOM's, in addition to the GVD. Since they > are > diffractive elements, they would introduce spatial chirp if used with > a > femtosecond beam with a large spectral content ( if we are not using > the zero > order). However, I have not properly tested them to see if this > really matters. > May be the spectral chirp is properly compensated by the AOM > manufacturers. Has > anyone in this group tested it for the spatial chirp? > > Sudipta > > > On Tue, 24 Sep 2019 06:03:38 +0000, Peter Müller wrote > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Dear Microscopists, > > for our multiphoton scope we use an acousto-optical modulator for > > intensity modulation. Those AOM have a nice overall throughput (>85%) > > and a high bandwidth (>1MHz), however they add a lot of group delay > > dispersion (GDD) to our setup - typically they use a 1cm - 2cm TeO2 > > crystal for the NIR wavelengths from 700nm to 1000nm. For our > > multiphoton setup, the AOM is the main contributor of GDD right now. > > I was wondering: what is your preferred choice for intensity > > modulation and what are the respective advantages/disadvantages? > > > > Thanks,Peter > > > Prof. Sudipta Maiti > Dept. of Chemical Sciences > Tata Institute of Fundamental Research > Homi Bhabha Road, Colaba > Mumbai 400005, India > Ph. +91 222 278 2716 > Alternate e-mail: [hidden email] > webpage: biophotonics.co.in > > > > -- -- Zdenek Svindrych, Ph.D. Research Associate - Imaging Specialist Department of Biochemistry and Cell Biology Geisel School of Medicine at Dartmouth |