Gary G. Li-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Greetings, For one photon excitation, the excitation and emission spectra of commonly used fluorescent dyes can be viewed by using the following tools: * Invitrogen Fluorescence SpectraViewer, www.invitrogen.com * BD Fluorescence Spectrum Viewer, www.bdbiosciences.com I am wondering if we have similar tools for fluorophores with multiphoton excitation. Many thanks Gary G. Li, Ph.D. Microscopy Core Manager NIH/NIDA 333 Cassell Drive, Suite 2100 Baltimore, MD 21224 [hidden email] [hidden email] <[hidden email]> <[hidden email]> |
|
Farid Jalali |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello Gary, I have found the DRBIO website (Developmental Resource for Biophysical Imaging Opto-Electronics at Cornell- Warren Zipfel, Watt Webb and crew ) at http://www.drbio.cornell.edu/cross_sections.html to be an excellent resource for at least the 2P excitation cross sections. The following paper is good commonly used reagents. I have yet to find anything equivalent to the resources you mention however. Two-photon fluorescence absorption and emission spectra of dyes relevant for cell imaging. J Microsc. 2002 Nov;208(Pt 2):108-15 Bestvater F, Spiess E, Stobrawa G, Hacker M, Feurer T, Porwol T, Berchner-Pfannschmidt U, Wotzlaw C, Acker H. Cheers Farid On Tue, Apr 5, 2011 at 8:03 AM, Gary G. Li <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Greetings, > > For one photon excitation, the excitation and emission spectra of commonly > used fluorescent dyes can be viewed by using the following tools: > > * Invitrogen Fluorescence SpectraViewer, www.invitrogen.com > > * BD Fluorescence Spectrum Viewer, www.bdbiosciences.com > > I am wondering if we have similar tools for fluorophores with multiphoton > excitation. > > Many thanks > > Gary G. Li, Ph.D. > > Microscopy Core Manager > > NIH/NIDA > > 333 Cassell Drive, Suite 2100 > > Baltimore, MD 21224 > > [hidden email] > > [hidden email] <[hidden email]> > > > <[hidden email]> > |
|
Johannes Helm |
|
|
In reply to this post by Gary G. Li-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Good afternoon, while I am not aware of a similar on line tool for multi photon excitations, there are publications, which list "MP" excitation spectra of a number of dyes, e.g. by Bestvater et al. Ref. Bestvater, F. et al. Two-photon fluorescence absorption and emission spectra of dyes relevant for cell imaging, Journal of Microscopy, Vol. 208, Pt 2 November 2002, pp. 108115 Best wishes, Johannes > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Greetings, > > For one photon excitation, the excitation and emission spectra of commonly > used fluorescent dyes can be viewed by using the following tools: > > * Invitrogen Fluorescence SpectraViewer, www.invitrogen.com > > * BD Fluorescence Spectrum Viewer, www.bdbiosciences.com > > I am wondering if we have similar tools for fluorophores with multiphoton > excitation. > > Many thanks > > Gary G. Li, Ph.D. > > Microscopy Core Manager > > NIH/NIDA > > 333 Cassell Drive, Suite 2100 > > Baltimore, MD 21224 > > [hidden email] > > [hidden email] <[hidden email]> > > > <[hidden email]> > -- P. Johannes Helm, M.Sc. PhD Seniorengineer CMBN University of Oslo Institute of Basic Medical Science Department of Anatomy Postboks 1105 - Blindern NO-0317 Oslo Voice: +47 228 51159 Fax: +47 228 51499 WWW: folk.uio.no/jhelm |
|
James Fitzpatrick |
|
|
In reply to this post by Gary G. Li-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Gary, Have you tried the DRBIO resource at Cornell that is run by Warren Zipfel now? They have a list of two-photon absorption cross-sections there that I find helpful. Its not exhaustive, but they have a number of the fluorescent proteins and alexa dyes listed there for example, and you can download the ascii data as well. http://www.drbio.cornell.edu/cross_sections.html Best wishes, James. On Apr 5, 2011, at 8:03 AM, Gary G. Li wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Greetings, > > For one photon excitation, the excitation and emission spectra of commonly > used fluorescent dyes can be viewed by using the following tools: > > * Invitrogen Fluorescence SpectraViewer, www.invitrogen.com > > * BD Fluorescence Spectrum Viewer, www.bdbiosciences.com > > I am wondering if we have similar tools for fluorophores with multiphoton > excitation. > > Many thanks > > Gary G. Li, Ph.D. > > Microscopy Core Manager > > NIH/NIDA > > 333 Cassell Drive, Suite 2100 > > Baltimore, MD 21224 > > [hidden email] > > [hidden email] <[hidden email]> > > > <[hidden email]> > ----- James A.J. Fitzpatrick, Ph.D. Core Director Waitt Advanced Biophotonics Center The Salk Institute for Biological Studies 10010 North Torrey Pines Road La Jolla, CA 92037 (t): +1 858-453-4100 x2119 (e): [hidden email] (w): http:/www.salk.edu/biophotonics_core |
|
Boswell, Carl A - (cboswell) |
|
|
In reply to this post by Farid Jalali
Probably the most comprehensive site would be at: http://www.spectra.arizona.edu/ . We've compiled all the 2P data we can find, including the DRBIO data. Simply select "2-photon cross-sections" under Dye category. We hope to incorporate the ability to download any data the user chooses, soon. This site has been assembled by Dr. Urs Utzinger at the U of A Biomedical Engineering Dept. and myself, with the code-writing help of Gary Carstensen in our Instruction/Access office, and includes much of the 1P data collected by Dr. George McNamara, U. of Miami, and myself.
Additional spectral data from optics, dyes, filters and light sources always welcome. Cheers, C Carl A. Boswell, Ph.D. Molecular and Cellular Biology Univ. of Arizona 520-954-7053 FAX 520-621-3709 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Farid Jalali Sent: Tuesday, April 05, 2011 8:30 AM To: [hidden email] Subject: Re: Multiphoton spectra of fluorescent dyes ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello Gary, I have found the DRBIO website (Developmental Resource for Biophysical Imaging Opto-Electronics at Cornell- Warren Zipfel, Watt Webb and crew ) at http://www.drbio.cornell.edu/cross_sections.html to be an excellent resource for at least the 2P excitation cross sections. The following paper is good commonly used reagents. I have yet to find anything equivalent to the resources you mention however. Two-photon fluorescence absorption and emission spectra of dyes relevant for cell imaging. J Microsc. 2002 Nov;208(Pt 2):108-15 Bestvater F, Spiess E, Stobrawa G, Hacker M, Feurer T, Porwol T, Berchner-Pfannschmidt U, Wotzlaw C, Acker H. Cheers Farid On Tue, Apr 5, 2011 at 8:03 AM, Gary G. Li <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Greetings, > > For one photon excitation, the excitation and emission spectra of > commonly used fluorescent dyes can be viewed by using the following tools: > > * Invitrogen Fluorescence SpectraViewer, www.invitrogen.com > > * BD Fluorescence Spectrum Viewer, www.bdbiosciences.com > > I am wondering if we have similar tools for fluorophores with > multiphoton excitation. > > Many thanks > > Gary G. Li, Ph.D. > > Microscopy Core Manager > > NIH/NIDA > > 333 Cassell Drive, Suite 2100 > > Baltimore, MD 21224 > > [hidden email] > > [hidden email] <[hidden email]> > > > <[hidden email]> > |
|
John Oreopoulos |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Two other publications of interest for you: Dickinson, M.E., et al., Multiphoton excitation spectra in biological samples. Journal of Biomedical Optics, 2003. 8(3): p. 329-338. Spiess, E., et al., Two-photon excitation and emission spectra of the green fluorescent protein variants ecfp, egfp and eyfp. Journal of Microscopy-Oxford, 2005. 217: p. 200-204. John Oreopoulos Research Assistant Spectral Applied Research Richmond Hill, Ontario Canada www.spectral.ca On 2011-04-05, at 7:46 PM, Boswell, Carl A - (cboswell) wrote: > Probably the most comprehensive site would be at: http://www.spectra.arizona.edu/ . We've compiled all the 2P data we can find, including the DRBIO data. Simply select "2-photon cross-sections" under Dye category. We hope to incorporate the ability to download any data the user chooses, soon. This site has been assembled by Dr. Urs Utzinger at the U of A Biomedical Engineering Dept. and myself, with the code-writing help of Gary Carstensen in our Instruction/Access office, and includes much of the 1P data collected by Dr. George McNamara, U. of Miami, and myself. > > Additional spectral data from optics, dyes, filters and light sources always welcome. > > Cheers, > C > > > > Carl A. Boswell, Ph.D. > Molecular and Cellular Biology > Univ. of Arizona > 520-954-7053 > FAX 520-621-3709 > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Farid Jalali > Sent: Tuesday, April 05, 2011 8:30 AM > To: [hidden email] > Subject: Re: Multiphoton spectra of fluorescent dyes > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello Gary, > > I have found the DRBIO website (Developmental Resource for Biophysical Imaging Opto-Electronics at Cornell- Warren Zipfel, Watt Webb and crew > ) at http://www.drbio.cornell.edu/cross_sections.html to be an excellent resource for at least the 2P excitation cross sections. The following paper is good commonly used reagents. I have yet to find anything equivalent to the resources you mention however. > > Two-photon fluorescence absorption and emission spectra of dyes relevant for cell imaging. > J Microsc. 2002 Nov;208(Pt 2):108-15 > Bestvater F, Spiess E, Stobrawa G, Hacker M, Feurer T, Porwol T, Berchner-Pfannschmidt U, Wotzlaw C, Acker H. > > > Cheers > Farid > > On Tue, Apr 5, 2011 at 8:03 AM, Gary G. Li <[hidden email]> wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Greetings, >> >> For one photon excitation, the excitation and emission spectra of >> commonly used fluorescent dyes can be viewed by using the following tools: >> >> * Invitrogen Fluorescence SpectraViewer, www.invitrogen.com >> >> * BD Fluorescence Spectrum Viewer, www.bdbiosciences.com >> >> I am wondering if we have similar tools for fluorophores with >> multiphoton excitation. >> >> Many thanks >> >> Gary G. Li, Ph.D. >> >> Microscopy Core Manager >> >> NIH/NIDA >> >> 333 Cassell Drive, Suite 2100 >> >> Baltimore, MD 21224 >> >> [hidden email] >> >> [hidden email] <[hidden email]> >> >> >> <[hidden email]> >> > |
|
Marco Dal Maschio |
|
|
In reply to this post by Gary G. Li-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** let me suggest for reddish emitting molecules the following paper where are resumed 9 crosssections including tdtomato, mcherry, mRFP and dsred2. Absolute two-photon absorption spectra and two-photon brightness of orange and red fluorescent proteins. <http://www.ncbi.nlm.nih.gov/pubmed/19127988> Drobizhev M, Tillo S, Makarov NS, Hughes TE, Rebane A. J Phys Chem B. 2009 Jan 29;113(4):855-9. My best Marco -- Marco Dal Maschio, PhD PostDoc fellow Neuroscience and Brain Technologies Dept Italian Institute of Technology via Morego 30, Genova, Italy |
|
Boswell, Carl A - (cboswell) |
|
|
These have been incorporated into the database.
C Carl A. Boswell, Ph.D. Molecular and Cellular Biology Univ. of Arizona 520-954-7053 FAX 520-621-3709 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Marco Dal Maschio Sent: Wednesday, April 06, 2011 12:50 AM To: [hidden email] Subject: Re: Multiphoton spectra of fluorescent dyes ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** let me suggest for reddish emitting molecules the following paper where are resumed 9 crosssections including tdtomato, mcherry, mRFP and dsred2. Absolute two-photon absorption spectra and two-photon brightness of orange and red fluorescent proteins. <http://www.ncbi.nlm.nih.gov/pubmed/19127988> Drobizhev M, Tillo S, Makarov NS, Hughes TE, Rebane A. J Phys Chem B. 2009 Jan 29;113(4):855-9. My best Marco -- Marco Dal Maschio, PhD PostDoc fellow Neuroscience and Brain Technologies Dept Italian Institute of Technology via Morego 30, Genova, Italy |