To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Fellow (confocal) microscopists, Excluding lasers, does anyone know of a light source that would emit in NIR (around 800nm) wavelength with a good quantum efficiency? Are there other benchmarks to measure signal power? Thank you Judy Judy Trogadis Bio-Imaging Coordinator Keenan Research Centre, St. Michael's 209 Victoria Street Toronto M5B 1T8, Canada office: 416-864-6060 ext. 77612 imaging facility: ext. 77434 cell: 416-909-9878 [hidden email]<mailto:[hidden email]> |
|
Martin Wessendorf-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Judy-- On 3/19/2012 12:30 PM, Judy Trogadis wrote: > Excluding lasers, does anyone know of a light source that would emit in NIR (around 800nm) wavelength with a good quantum efficiency? Are there other benchmarks to measure signal power? A conventional incandescent bulb is pretty good--I've used one for NIR DIC--but it really depends upon how much power you need. The following link (no commercial connection) lists NIR LEDs in the 800 nm range; no idea of cost: http://www.roithner-laser.com/led_diverse.html Good luck! Martin -- Martin Wessendorf, Ph.D. office: (612) 626-0145 Assoc Prof, Dept Neuroscience lab: (612) 624-2991 University of Minnesota Preferred FAX: (612) 624-8118 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 Minneapolis, MN 55455 e-mail: [hidden email] |
|
|
In reply to this post by Judy Trogadis-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Try this http://www.thorlabs.com/NewGroupPage9.cfm?ObjectGroup_ID=2615 |
|
Judy Trogadis-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi, Pascal, These look amazingly practical. Too good to be true? Will be talking to Thorlabs. Thank you! Judy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Pascal Weber Sent: Monday, March 19, 2012 4:15 PM To: [hidden email] Subject: Re: NIR excitation light source ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Try this http://www.thorlabs.com/NewGroupPage9.cfm?ObjectGroup_ID=2615 |
|
Judy Trogadis-2 |
|
|
In reply to this post by Martin Wessendorf-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello, Martin In addition to having a drawer of filter cubes, we will now have a drawer of LED's. Not a bad solution. Thank you, the prices are excellent but I need an electronics expert because these are just the raw components. Perhaps my nephew in his basement . . . . . . . Judy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Martin Wessendorf Sent: Monday, March 19, 2012 1:46 PM To: [hidden email] Subject: Re: NIR excitation light source ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Judy-- On 3/19/2012 12:30 PM, Judy Trogadis wrote: > Excluding lasers, does anyone know of a light source that would emit in NIR (around 800nm) wavelength with a good quantum efficiency? Are there other benchmarks to measure signal power? A conventional incandescent bulb is pretty good--I've used one for NIR DIC--but it really depends upon how much power you need. The following link (no commercial connection) lists NIR LEDs in the 800 nm range; no idea of cost: http://www.roithner-laser.com/led_diverse.html Good luck! Martin -- Martin Wessendorf, Ph.D. office: (612) 626-0145 Assoc Prof, Dept Neuroscience lab: (612) 624-2991 University of Minnesota Preferred FAX: (612) 624-8118 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 Minneapolis, MN 55455 e-mail: [hidden email] |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** To drive an LED you need a 'constant current' supply. You can make simple ones with a small part called an LM317, but if you aren't up to electronics hacking then I'd recommend getting one of Thor's T-Cube drivers. The LEDD1B is just under $300 USD and will drive an LED up to 1200mA. It's not a bad deal and it should already be wired for the connectors of Thor's own LED products. I think it would save you a lot of time and headache. http://www.thorlabs.com/NewGroupPage9.cfm?ObjectGroup_ID=2616 Craig On Mon, Mar 19, 2012 at 2:40 PM, Judy Trogadis <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello, Martin > > In addition to having a drawer of filter cubes, we will now have a drawer > of LED's. Not a bad solution. Thank you, the prices are excellent but I > need an electronics expert because these are just the raw components. > Perhaps my nephew in his basement . . . . . . . > > Judy > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Martin Wessendorf > Sent: Monday, March 19, 2012 1:46 PM > To: [hidden email] > Subject: Re: NIR excitation light source > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Judy-- > > On 3/19/2012 12:30 PM, Judy Trogadis wrote: > > > Excluding lasers, does anyone know of a light source that would emit in > NIR (around 800nm) wavelength with a good quantum efficiency? Are there > other benchmarks to measure signal power? > > A conventional incandescent bulb is pretty good--I've used one for NIR > DIC--but it really depends upon how much power you need. > > The following link (no commercial connection) lists NIR LEDs in the 800 > nm range; no idea of cost: http://www.roithner-laser.com/led_diverse.html > > Good luck! > > Martin > -- > Martin Wessendorf, Ph.D. office: (612) 626-0145 > Assoc Prof, Dept Neuroscience lab: (612) 624-2991 > University of Minnesota Preferred FAX: (612) 624-8118 > 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 > Minneapolis, MN 55455 e-mail: [hidden email] > |
|
|
In reply to this post by Judy Trogadis-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I have some questions for you. What do you do exactly. Which microscope would you adapt your installation? We found on the internet diodes more powerful than 10 watts. |
|
Judy Trogadis-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Whole animal imagers often use NIR dyes for a variety of different reasons. After processing thin sections for histology, we would like to image those same areas. From the Thorlabs website, it seems as if we can simply exchange the light guide in the epillumination port on one of our microscopes (acceptable models are also listed) with the LED's attachment and it's just a matter of matching the light with the appropriate filter cube. I think 10 watts would provide enough light, don't you think? Judy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Pascal Weber Sent: Monday, March 19, 2012 5:19 PM To: [hidden email] Subject: Re: NIR excitation light source ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I have some questions for you. What do you do exactly. Which microscope would you adapt your installation? We found on the internet diodes more powerful than 10 watts. |
|
Judy Trogadis-2 |
|
|
In reply to this post by Craig Brideau
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** You're the second person who has recommended Thorlabs. You guys know it all! Thank you! Judy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau Sent: Monday, March 19, 2012 4:57 PM To: [hidden email] Subject: Re: NIR excitation light source ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** To drive an LED you need a 'constant current' supply. You can make simple ones with a small part called an LM317, but if you aren't up to electronics hacking then I'd recommend getting one of Thor's T-Cube drivers. The LEDD1B is just under $300 USD and will drive an LED up to 1200mA. It's not a bad deal and it should already be wired for the connectors of Thor's own LED products. I think it would save you a lot of time and headache. http://www.thorlabs.com/NewGroupPage9.cfm?ObjectGroup_ID=2616 Craig On Mon, Mar 19, 2012 at 2:40 PM, Judy Trogadis <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello, Martin > > In addition to having a drawer of filter cubes, we will now have a drawer > of LED's. Not a bad solution. Thank you, the prices are excellent but I > need an electronics expert because these are just the raw components. > Perhaps my nephew in his basement . . . . . . . > > Judy > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Martin Wessendorf > Sent: Monday, March 19, 2012 1:46 PM > To: [hidden email] > Subject: Re: NIR excitation light source > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Judy-- > > On 3/19/2012 12:30 PM, Judy Trogadis wrote: > > > Excluding lasers, does anyone know of a light source that would emit in > NIR (around 800nm) wavelength with a good quantum efficiency? Are there > other benchmarks to measure signal power? > > A conventional incandescent bulb is pretty good--I've used one for NIR > DIC--but it really depends upon how much power you need. > > The following link (no commercial connection) lists NIR LEDs in the 800 > nm range; no idea of cost: http://www.roithner-laser.com/led_diverse.html > > Good luck! > > Martin > -- > Martin Wessendorf, Ph.D. office: (612) 626-0145 > Assoc Prof, Dept Neuroscience lab: (612) 624-2991 > University of Minnesota Preferred FAX: (612) 624-8118 > 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 > Minneapolis, MN 55455 e-mail: [hidden email] > |
|
Mike Buchin |
|
|
In reply to this post by Pascal Weber
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We have had good luck with Prizmatix www.prizmatix.com high power LED'.s They only show product out to 630 nm but I am sure they could provide NIR (730, 850 etc.). Their drivers are fast (several microseconds) switching and high bandwidth (up to 30 kHz rep rates). Direct or fiber coupled to all scope configurations. No commercial interest. Mike Michael Buchin Stanford Photonics, Inc. Ph: 650-969-5991 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Pascal Weber Sent: Monday, March 19, 2012 2:19 PM To: [hidden email] Subject: Re: NIR excitation light source ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I have some questions for you. What do you do exactly. Which microscope would you adapt your installation? We found on the internet diodes more powerful than 10 watts. |
|
|
In reply to this post by Judy Trogadis-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Do you really think that these fluorophores emit around 800nm ​? It is always possible to adapt the cubes. Will see in SEMROCK. http://www.semrock.com/filters.aspx Ten watts should be enough but yo need to do electronics. are you a specialist ? |
|
Judy Trogadis-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Yes, they do - one in particular - ICG (indocyanine green) and filter companies sell cubes specifically for that dye. Emission around 810nm. Judy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Pascal Weber Sent: Monday, March 19, 2012 5:40 PM To: [hidden email] Subject: Re: NIR excitation light source ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Do you really think that these fluorophores emit around 800nm ​? It is always possible to adapt the cubes. Will see in SEMROCK. http://www.semrock.com/filters.aspx Ten watts should be enough but yo need to do electronics. are you a specialist ? |
|
Pascal Weber |
|
|
In reply to this post by Mike Buchin
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** The excitation wavelengths are 660 and 780 nm. It is not exactly the same. You will find the dichroic and emission filter at Semrock or Chroma. To measure the power I use a mirror purchased from Thorlabs and compares two wavelengths from the visible and the one you just installed. |
|
|
In reply to this post by Judy Trogadis-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** See that http://www.alibaba.com/product- gs/493247250/660nm_10w_red_led_professional_manufacturer_.html It's a beginning............. |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Heh, yes, but just make sure you properly heat sink it! Otherwise you will have brilliant light for about a minute, then it will go from being an LED to a DED... Craig On Mon, Mar 19, 2012 at 4:13 PM, Pascal Weber <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > See that > > http://www.alibaba.com/product- > gs/493247250/660nm_10w_red_led_professional_manufacturer_.html > > It's a beginning............. > |