NIR lamp source

classic Classic list List threaded Threaded
♦
♦
4 messages Options
Claudio Vinegoni Claudio Vinegoni
Reply | Threaded
Open this post in threaded view
♦
♦
|

NIR lamp source

Does anyone know of a good lamp for fluorescence microscopy that covers
visible and near infrared (up to 750 nm)? Or alternatively a lamp that could
be used in the range 600 up to 750?
Checked Sutter and Exfo but they die pretty quickly at around 600
Thansk for any suggestion
            -Claudio
Markus Neumann Markus Neumann
Reply | Threaded
Open this post in threaded view
♦
♦
|

AW: NIR lamp source

Why don't you try the good old Halogen lamp? Its spectrum in the IR is pretty good, better than most arc lamps. If you take out the heat filter (which is what removes the strong IR part of the spectrum), you will have plenty of IR. You can check out the spectrum here: http://zeiss-campus.magnet.fsu.edu/articles/lightsources/tungstenhalogen.html

Regards

Markus

-----Ursprüngliche Nachricht-----
Von: Confocal Microscopy List [mailto:[hidden email]] Im Auftrag von Claudio Vinegoni
Gesendet: Freitag, 14. November 2008 16:40
An: [hidden email]
Betreff: NIR lamp source

Does anyone know of a good lamp for fluorescence microscopy that covers
visible and near infrared (up to 750 nm)? Or alternatively a lamp that could
be used in the range 600 up to 750?
Checked Sutter and Exfo but they die pretty quickly at around 600
Thansk for any suggestion
            -Claudio
Claudio Vinegoni Claudio Vinegoni
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: AW: NIR lamp source

Dear Markus
thanks a lot for your comment. I've checked before and there is a good amount of signal at 750 with one of our halogen. I was going to try a led too. I'll see what will work better
thanks again
          -Claudio

***************************************************

Claudio Vinegoni, Ph.D.

Center for System Biology
MGH-Harvard University
Richard B. Simches Research Center

185 Cambridge Street
8th Floor CPZN5
Boston, MA 02114
Ph: 857-891-4272
F: 617-726-5708

http://csb.mgh.harvard.edu/
http://www.mgh-cmir.org/
http://WWW.QUANTACOMM.COM/

***************************************************

--- On Fri, 11/14/08, Markus Neumann <[hidden email]> wrote:
From: Markus Neumann <[hidden email]>
Subject: AW: NIR lamp source
To: [hidden email]
Date: Friday, November 14, 2008, 8:09 PM

Why don't you try the good old Halogen lamp? Its spectrum in the IR is
pretty good, better than most arc lamps. If you take out the heat filter (which
is what removes the strong IR part of the spectrum), you will have plenty of IR.
You can check out the spectrum here:
http://zeiss-campus.magnet.fsu.edu/articles/lightsources/tungstenhalogen.html

Regards

Markus

-----Ursprüngliche Nachricht-----
Von: Confocal Microscopy List [mailto:[hidden email]] Im
Auftrag von Claudio Vinegoni
Gesendet: Freitag, 14. November 2008 16:40
An: [hidden email]
Betreff: NIR lamp source

Does anyone know of a good lamp for fluorescence microscopy that covers
visible and near infrared (up to 750 nm)? Or alternatively a lamp that could
be used in the range 600 up to 750?
Checked Sutter and Exfo but they die pretty quickly at around 600 
Thansk for any suggestion
            -Claudio

Peter Gabriel Pitrone Peter Gabriel Pitrone
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: NIR lamp source

In reply to this post by Claudio Vinegoni
Xeon


On Nov 14, 2008, at 4:40 PM, Claudio Vinegoni wrote:

> Does anyone know of a good lamp for fluorescence microscopy that  
> covers
> visible and near infrared (up to 750 nm)? Or alternatively a lamp  
> that could
> be used in the range 600 up to 750?
> Checked Sutter and Exfo but they die pretty quickly at around 600
> Thansk for any suggestion
>             -Claudio

Peter Gabriel Pitrone
Light Microscopy Facility   (LMF)
Max Planck Institute for Molecular
Cell Biology and Genetics (MPI-CBG)
Pfotenhauerstrasse 108
01307 Dresden, Germany

Tele: +49 (0)351-210-2687
E-Mail: [hidden email]