![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi list. What is the proper name of this connection? https://photos.app.goo.gl/UmsycFt2qKK3W94g6 Thanks. Petro. |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** C-mount??? -----Mensaje original----- De: Confocal Microscopy List [mailto:[hidden email]] En nombre de Piter_ Enviado el: lunes, 11 de marzo de 2019 15:17 Para: [hidden email] Asunto: Name of connection ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi list. What is the proper name of this connection? https://photos.app.goo.gl/UmsycFt2qKK3W94g6 Thanks. Petro. |
![]() |
In reply to this post by Piter_
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Isn't that just the camera port? On Mon, Mar 11, 2019 at 11:18 AM Piter_ <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi list. > What is the proper name of this connection? > https://photos.app.goo.gl/UmsycFt2qKK3W94g6 > Thanks. > Petro. > -- Benjamin E. Smith, Ph. D. Imaging Specialist, Vision Science University of California, Berkeley 195 Life Sciences Addition Berkeley, CA 94720-3200 Tel (510) 642-9712 Fax (510) 643-6791 e-mail: [hidden email] http://vision.berkeley.edu/?page_id=5635 <http://vision.berkeley.edu/> |
![]() |
In reply to this post by Piter_
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Female Dovetail: https://www.thorlabs.com/newgrouppage9.cfm?objectgroup_id=8563&TabName=Microscope%20Dovetails Mike On Mon, Mar 11, 2019 at 2:19 PM Piter_ <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi list. > What is the proper name of this connection? > https://photos.app.goo.gl/UmsycFt2qKK3W94g6 > Thanks. > Petro. |
![]() |
In reply to this post by Piter_
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Petro, Are you referring to the C-mount? For connecting camera to microscope and specific to the microscope? Cheers, Jacqui ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of Piter_ [[hidden email]] Sent: 12 March 2019 07:17 To: [hidden email] Subject: Name of connection ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi list. What is the proper name of this connection? https://photos.app.goo.gl/UmsycFt2qKK3W94g6 Thanks. Petro. |
![]() ![]() |
Schwartz, Owen (NIH/NIAID) [E]-2 |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I would call this a photo port. The C mount is the adapter with the flange at one end and the thread at the other for connecting the camera. I believe in fact that the "C mount" is referring to the thread mount style of the camera. Be aware that the flanges on the microscope come in different diameters depending on brand. In addition the C Mount adapters can be 0.5x, 0.7x, 1x, etc. depending on the size of the camera chip. Cheers, Owen Owen M. Schwartz Head, Biological Imaging Facility Senior Associate Scientist NIAID-NIH Bldg 4 rm B2-30 4 Center Drive Bethesda, MD 20892 Phone 240-669-5442 Fax 301-280-2882 -----Original Message----- From: Jacqueline Ross <[hidden email]> Sent: Monday, March 11, 2019 4:31 PM To: [hidden email] Subject: Re: Name of connection ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Petro, Are you referring to the C-mount? For connecting camera to microscope and specific to the microscope? Cheers, Jacqui ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of Piter_ [[hidden email]] Sent: 12 March 2019 07:17 To: [hidden email] Subject: Name of connection ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi list. What is the proper name of this connection? https://photos.app.goo.gl/UmsycFt2qKK3W94g6 Thanks. Petro. |
![]() |
In reply to this post by Jacqueline Ross
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** C-mount is a threaded connection, typically used for video lenses mounted right before the CCD/CMOS array. The port in the picture is a dovetail, as Michael said, specs of which depend on the microscope brand. Craig On Mon, Mar 11, 2019 at 2:31 PM Jacqueline Ross <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Petro, > > Are you referring to the C-mount? For connecting camera to microscope and > specific to the microscope? > > Cheers, > > Jacqui > ________________________________________ > From: Confocal Microscopy List [[hidden email]] on > behalf of Piter_ [[hidden email]] > Sent: 12 March 2019 07:17 > To: [hidden email] > Subject: Name of connection > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi list. > What is the proper name of this connection? > https://photos.app.goo.gl/UmsycFt2qKK3W94g6 > Thanks. > Petro. > |
![]() |
In reply to this post by Schwartz, Owen (NIH/NIAID) [E]-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi everybody. Thanks for replies. It looks like a proper name is a photo port, with a circular dovetail. I was hoping that it is standardized. I have a Leica DMIRE2, and building a palm PALM setup on it. But with my objectives and camera will need >2X magnification changer between the mic and a c-mount camera. I spoke with a local Leica representative a week ago and he told me that they don't have it. So my hope was to get specifications for the dovetail connections and make cage adapter in a workshop. Then I can combine some lenses. May be somebody have seen somewhere online 2x c-mount adapter for this port, Thank you Petro. On Mon, Mar 11, 2019 at 9:53 PM Schwartz, Owen (NIH/NIAID) [E] < [hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I would call this a photo port. The C mount is the adapter with the > flange at one end and the thread at the other for connecting the camera. I > believe in fact that the "C mount" is referring to the thread mount style > of the camera. Be aware that the flanges on the microscope come in > different diameters depending on brand. In addition the C Mount adapters > can be 0.5x, 0.7x, 1x, etc. depending on the size of the camera chip. > > Cheers, > > Owen > > > > Owen M. Schwartz > Head, Biological Imaging Facility > Senior Associate Scientist > NIAID-NIH > Bldg 4 rm B2-30 > 4 Center Drive > Bethesda, MD 20892 > Phone 240-669-5442 > Fax 301-280-2882 > > -----Original Message----- > From: Jacqueline Ross <[hidden email]> > Sent: Monday, March 11, 2019 4:31 PM > To: [hidden email] > Subject: Re: Name of connection > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Petro, > > Are you referring to the C-mount? For connecting camera to microscope and > specific to the microscope? > > Cheers, > > Jacqui > ________________________________________ > From: Confocal Microscopy List [[hidden email]] on > behalf of Piter_ [[hidden email]] > Sent: 12 March 2019 07:17 > To: [hidden email] > Subject: Name of connection > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi list. > What is the proper name of this connection? > https://photos.app.goo.gl/UmsycFt2qKK3W94g6 > Thanks. > Petro. > |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Piter, I can't tell from that picture exactly which model of dovetail you have, but it should be easy to measure. Assuming it isn't something very unusual, you can buy adapters online once you know which dovetail you have: https://www.thorlabs.com/newgrouppage9.cfm?objectgroup_id=10093 Mike On Mon, Mar 11, 2019 at 6:33 PM Piter_ <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIBaQ&c=kbmfwr1Yojg42sGEpaQh5ofMHBeTl9EI2eaqQZhHbOU&r=0LyF_z8oU1XGGyisIeOIXyIGIM5IYb3NcLjxHjUca5Y&m=x0u9Mb4MyQFlhMoKfJYd6I2tNoIdvHo7YCBTlbbsm_k&s=f7jd_TKajPqf1meyjdHXs019eFew6szPBfLwyBoOKPM&e= > Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIBaQ&c=kbmfwr1Yojg42sGEpaQh5ofMHBeTl9EI2eaqQZhHbOU&r=0LyF_z8oU1XGGyisIeOIXyIGIM5IYb3NcLjxHjUca5Y&m=x0u9Mb4MyQFlhMoKfJYd6I2tNoIdvHo7YCBTlbbsm_k&s=LKmNzsUTU6YMCyxqBZgvvh0plPUXbjPqpgdY1r1kipw&e= and include the link in your posting. > ***** > > Hi everybody. > Thanks for replies. > It looks like a proper name is a photo port, with a circular dovetail. I > was hoping that it is standardized. > I have a Leica DMIRE2, and building a palm PALM setup on it. But with my > objectives and camera will need >2X magnification changer between the mic > and a c-mount camera. I spoke with a local Leica representative a week ago > and he told me that they don't have it. So my hope was to get > specifications for the dovetail connections and make cage adapter in a > workshop. Then I can combine some lenses. > May be somebody have seen somewhere online 2x c-mount adapter for this port, > Thank you > Petro. > > On Mon, Mar 11, 2019 at 9:53 PM Schwartz, Owen (NIH/NIAID) [E] < > [hidden email]> wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIBaQ&c=kbmfwr1Yojg42sGEpaQh5ofMHBeTl9EI2eaqQZhHbOU&r=0LyF_z8oU1XGGyisIeOIXyIGIM5IYb3NcLjxHjUca5Y&m=x0u9Mb4MyQFlhMoKfJYd6I2tNoIdvHo7YCBTlbbsm_k&s=f7jd_TKajPqf1meyjdHXs019eFew6szPBfLwyBoOKPM&e= > > Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIBaQ&c=kbmfwr1Yojg42sGEpaQh5ofMHBeTl9EI2eaqQZhHbOU&r=0LyF_z8oU1XGGyisIeOIXyIGIM5IYb3NcLjxHjUca5Y&m=x0u9Mb4MyQFlhMoKfJYd6I2tNoIdvHo7YCBTlbbsm_k&s=LKmNzsUTU6YMCyxqBZgvvh0plPUXbjPqpgdY1r1kipw&e= and include the link in your posting. > > ***** > > > > I would call this a photo port. The C mount is the adapter with the > > flange at one end and the thread at the other for connecting the camera. I > > believe in fact that the "C mount" is referring to the thread mount style > > of the camera. Be aware that the flanges on the microscope come in > > different diameters depending on brand. In addition the C Mount adapters > > can be 0.5x, 0.7x, 1x, etc. depending on the size of the camera chip. > > > > Cheers, > > > > Owen > > > > > > > > Owen M. Schwartz > > Head, Biological Imaging Facility > > Senior Associate Scientist > > NIAID-NIH > > Bldg 4 rm B2-30 > > 4 Center Drive > > Bethesda, MD 20892 > > Phone 240-669-5442 > > Fax 301-280-2882 > > > > -----Original Message----- > > From: Jacqueline Ross <[hidden email]> > > Sent: Monday, March 11, 2019 4:31 PM > > To: [hidden email] > > Subject: Re: Name of connection > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIBaQ&c=kbmfwr1Yojg42sGEpaQh5ofMHBeTl9EI2eaqQZhHbOU&r=0LyF_z8oU1XGGyisIeOIXyIGIM5IYb3NcLjxHjUca5Y&m=x0u9Mb4MyQFlhMoKfJYd6I2tNoIdvHo7YCBTlbbsm_k&s=f7jd_TKajPqf1meyjdHXs019eFew6szPBfLwyBoOKPM&e= > > Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIBaQ&c=kbmfwr1Yojg42sGEpaQh5ofMHBeTl9EI2eaqQZhHbOU&r=0LyF_z8oU1XGGyisIeOIXyIGIM5IYb3NcLjxHjUca5Y&m=x0u9Mb4MyQFlhMoKfJYd6I2tNoIdvHo7YCBTlbbsm_k&s=LKmNzsUTU6YMCyxqBZgvvh0plPUXbjPqpgdY1r1kipw&e= and include the link in your posting. > > ***** > > > > Hi Petro, > > > > Are you referring to the C-mount? For connecting camera to microscope and > > specific to the microscope? > > > > Cheers, > > > > Jacqui > > ________________________________________ > > From: Confocal Microscopy List [[hidden email]] on > > behalf of Piter_ [[hidden email]] > > Sent: 12 March 2019 07:17 > > To: [hidden email] > > Subject: Name of connection > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=DwIBaQ&c=kbmfwr1Yojg42sGEpaQh5ofMHBeTl9EI2eaqQZhHbOU&r=0LyF_z8oU1XGGyisIeOIXyIGIM5IYb3NcLjxHjUca5Y&m=x0u9Mb4MyQFlhMoKfJYd6I2tNoIdvHo7YCBTlbbsm_k&s=f7jd_TKajPqf1meyjdHXs019eFew6szPBfLwyBoOKPM&e= > > Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=DwIBaQ&c=kbmfwr1Yojg42sGEpaQh5ofMHBeTl9EI2eaqQZhHbOU&r=0LyF_z8oU1XGGyisIeOIXyIGIM5IYb3NcLjxHjUca5Y&m=x0u9Mb4MyQFlhMoKfJYd6I2tNoIdvHo7YCBTlbbsm_k&s=LKmNzsUTU6YMCyxqBZgvvh0plPUXbjPqpgdY1r1kipw&e= and include the link in your posting. > > ***** > > > > Hi list. > > What is the proper name of this connection? > > https://urldefense.proofpoint.com/v2/url?u=https-3A__photos.app.goo.gl_UmsycFt2qKK3W94g6&d=DwIBaQ&c=kbmfwr1Yojg42sGEpaQh5ofMHBeTl9EI2eaqQZhHbOU&r=0LyF_z8oU1XGGyisIeOIXyIGIM5IYb3NcLjxHjUca5Y&m=x0u9Mb4MyQFlhMoKfJYd6I2tNoIdvHo7YCBTlbbsm_k&s=hk-dVcZwJzp-e09shTMVCgk5Em_0ZioAHLlHqkrDjnI&e= > > Thanks. > > Petro. > > |
![]() |
In reply to this post by José M. Pellegrino
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Overheard in a pub: "I love you Americans and your standards; you have so many to choose from!" Microscopy in general seems to have the same issue... Craig On Mon, Mar 11, 2019, 6:41 PM Jose Pellegrino <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > C-mount??? > > -----Mensaje original----- > De: Confocal Microscopy List [mailto:[hidden email]] En > nombre de Piter_ > Enviado el: lunes, 11 de marzo de 2019 15:17 > Para: [hidden email] > Asunto: Name of connection > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi list. > What is the proper name of this connection? > https://photos.app.goo.gl/UmsycFt2qKK3W94g6 > Thanks. > Petro. > |
![]() |
In reply to this post by Piter_
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Petro, that's is a TV/camera port, You need a C mount adapter like the ones reported at the link https://drive.google.com/open?id=1u42itZxmUSgWkuuWe8jJ_N5SanCLRZ0k The 1X should work in most of the cases and yet available (11541510) Regards Tommaso |
![]() ![]() |
Garner Oliver, BergmanLabora AB |
![]() |
In reply to this post by Piter_
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi again, Here's what you need (no commercial interest): https://webstore.diaginc.com/DE20HCC-2-0X-Adapter-p/wsn-de20hcc.htm Vänlig hälsning / Best regards Oliver Garner -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of Piter_ Sent: Monday, March 11, 2019 11:08 PM To: [hidden email] Subject: Re: Name of connection ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi everybody. Thanks for replies. It looks like a proper name is a photo port, with a circular dovetail. I was hoping that it is standardized. I have a Leica DMIRE2, and building a palm PALM setup on it. But with my objectives and camera will need >2X magnification changer between the mic and a c-mount camera. I spoke with a local Leica representative a week ago and he told me that they don't have it. So my hope was to get specifications for the dovetail connections and make cage adapter in a workshop. Then I can combine some lenses. May be somebody have seen somewhere online 2x c-mount adapter for this port, Thank you Petro. On Mon, Mar 11, 2019 at 9:53 PM Schwartz, Owen (NIH/NIAID) [E] < [hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I would call this a photo port. The C mount is the adapter with the > flange at one end and the thread at the other for connecting the > camera. I believe in fact that the "C mount" is referring to the > thread mount style of the camera. Be aware that the flanges on the > microscope come in different diameters depending on brand. In > addition the C Mount adapters can be 0.5x, 0.7x, 1x, etc. depending on the size of the camera chip. > > Cheers, > > Owen > > > > Owen M. Schwartz > Head, Biological Imaging Facility > Senior Associate Scientist > NIAID-NIH > Bldg 4 rm B2-30 > 4 Center Drive > Bethesda, MD 20892 > Phone 240-669-5442 > Fax 301-280-2882 > > -----Original Message----- > From: Jacqueline Ross <[hidden email]> > Sent: Monday, March 11, 2019 4:31 PM > To: [hidden email] > Subject: Re: Name of connection > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Petro, > > Are you referring to the C-mount? For connecting camera to microscope > and specific to the microscope? > > Cheers, > > Jacqui > ________________________________________ > From: Confocal Microscopy List [[hidden email]] on > behalf of Piter_ [[hidden email]] > Sent: 12 March 2019 07:17 > To: [hidden email] > Subject: Name of connection > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi list. > What is the proper name of this connection? > https://photos.app.goo.gl/UmsycFt2qKK3W94g6 > Thanks. > Petro. > |
![]() |
In reply to this post by Piter_
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Another model here which can work for you (no commercial interest): https://www.bestscientificcouplers.com/products/detail/coupler-links-1-c-mount-camera-leica-34.5mm-dovetail-interface1 You can use the coupler finder (I selected c-mount, 2x and Leica): https://www.bestscientificcouplers.com/products/23cc13cf3f16c73fcd0e6c748fe4c1de/74022b4bcc286d25e3d1ec4a22c94a1f Best, Vincent -----Original Message----- From: Confocal Microscopy List <[hidden email]> On Behalf Of Piter_ Sent: Monday, March 11, 2019 11:08 PM To: [hidden email] Subject: Re: Name of connection ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi everybody. Thanks for replies. It looks like a proper name is a photo port, with a circular dovetail. I was hoping that it is standardized. I have a Leica DMIRE2, and building a palm PALM setup on it. But with my objectives and camera will need >2X magnification changer between the mic and a c-mount camera. I spoke with a local Leica representative a week ago and he told me that they don't have it. So my hope was to get specifications for the dovetail connections and make cage adapter in a workshop. Then I can combine some lenses. May be somebody have seen somewhere online 2x c-mount adapter for this port, Thank you Petro. |
Free forum by Nabble | Edit this page |