|
Andrew Vaughan |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear all, I wonder if anyone out there can help me with a problem I'm having with the Perfect Focus System on a Nikon Ti. Basically, the PFS won't work when I use my 100X VC Plan Apochromat objective. My specimens are in various vessels, but all of them have coverslip thickness glass and at least 3mm of water/medium. When I focus on a specimen the PFS focus indicator light on the front of the stand indicates that the objective lens is not in the focusable range, even when the objective is focussed on the very surface of the coverslip. To test this I drew crosses in fine black marker pen on the inner coverslip surface of various chamber types. I then filled the chambers with water and focussed on the cross using transmitted light. My results are as follows: 1. The failure of the PFS seems to be limited to the 100X VC Plan Apochromat. Lower power lenses have consistently worked when tested. 2. If I move the stage so that the entire field is filled with the dark marker pen, the focus indicator lights up and PFS works. This is true whether or not the transmitted light is on or switched off. 3. If I get the PFS to work in the manner above and then move the stage so that Im looking at an area free of marker pen the PFS stays on unless I turn the transmitted light off or put an IR blocking filter in front of the transmitted light. 4. If I blindly (with the transmitted light off) move the stage back into an area where there is marker pen, the focus indicator light comes back on and the PFS will work. 5. Additionally, if I get the PFS working using the method above and then move the stage into a clear area while still observing using transmitted light and then try to adjust the offset using the PFS Offset Controller, this causes the PFS to fail again. From this I can say certain things: 1. Only the 100X VC Plan Apochromat is affected 2. A dark field (or possibly one with high contrast edges) is required for the PFS to work initially 3. The PFS can be sustained by IR light coming from the halogen lamp, but this isn't enough to get it going initially 4. Adjusting the offset, which moves the dichroic mirror in the PFS nosepiece, seems to take the system out of the focusable range Note that I cant rule out a fault with the objective, but I have compared this objective with another identical objective in other tests recently and have no reason to believe it is faulty. Any ideas or suggestions would be most welcome. Thanks, Andrew. |
|
Mario Faretta |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Andrew, we noticed the same behaviour also with other objectives and the hopinion I have is that the IR contribution from the transmitted light avoid the sensor (that is in the far red region) to read correctly the reflection from the glass. The reason to observe it only at high magnification is probably connected to the increase in excitation that you have to apply to the transmitted light source the improve the contrast. we solved it by blocking IR light or by reducing transmission adjusting exposure time on the camera (consequently not working at the ocular). You can also alternatively disactivate PFS from the software, acquire transmitted light image, block transmission and reactivate PFS. Hope it helps Mario Il 2/28/2012 2:07 PM, Andrew Vaughan ha scritto: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all, > > I wonder if anyone out there can help me with a problem I'm having with the > Perfect Focus System on a Nikon Ti. Basically, the PFS won't work when I use > my 100X VC Plan Apochromat objective. My specimens are in various vessels, > but all of them have coverslip thickness glass and at least 3mm of > water/medium. When I focus on a specimen the PFS focus indicator light on > the front of the stand indicates that the objective lens is not in the focusable > range, even when the objective is focussed on the very surface of the > coverslip. To test this I drew crosses in fine black marker pen on the inner > coverslip surface of various chamber types. I then filled the chambers with > water and focussed on the cross using transmitted light. My results are as > follows: > > 1. The failure of the PFS seems to be limited to the 100X VC Plan Apochromat. > Lower power lenses have consistently worked when tested. > 2. If I move the stage so that the entire field is filled with the dark marker > pen, the focus indicator lights up and PFS works. This is true whether or not > the transmitted light is on or switched off. > 3. If I get the PFS to work in the manner above and then move the stage so > that Im looking at an area free of marker pen the PFS stays on unless I turn > the transmitted light off or put an IR blocking filter in front of the transmitted > light. > 4. If I blindly (with the transmitted light off) move the stage back into an area > where there is marker pen, the focus indicator light comes back on and the > PFS will work. > 5. Additionally, if I get the PFS working using the method above and then > move the stage into a clear area while still observing using transmitted light > and then try to adjust the offset using the PFS Offset Controller, this causes > the PFS to fail again. > > From this I can say certain things: > > 1. Only the 100X VC Plan Apochromat is affected > 2. A dark field (or possibly one with high contrast edges) is required for the > PFS to work initially > 3. The PFS can be sustained by IR light coming from the halogen lamp, but > this isn't enough to get it going initially > 4. Adjusting the offset, which moves the dichroic mirror in the PFS nosepiece, > seems to take the system out of the focusable range > > Note that I cant rule out a fault with the objective, but I have compared this > objective with another identical objective in other tests recently and have no > reason to believe it is faulty. > > Any ideas or suggestions would be most welcome. > > Thanks, > > Andrew. -- ---PLEASE Note the change in telephone number--- -- Mario Faretta Department of Experimental Oncology European Institute of Oncology c/o IFOM-IEO Campus for Oncogenomics via Adamello 16 20139 Milan Italy Phone: ++39-0294375027 email: [hidden email] http://www.ifom-ieo-campus.it |
|
Kurt Thorn |
|
|
In reply to this post by Andrew Vaughan
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** You should really have an IR blocking filter in the transmitted light path. I suspect the IR from the halogen lamp is fooling the PFS into thinking it's seeing the coverslip reflection when really it's just seeing the IR from the transmitted light. At least in our scopes, Nikon always puts an IR blocking filter in the transmitted light path to prevent these problems. Kurt On 2/28/2012 5:07 AM, Andrew Vaughan wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all, > > I wonder if anyone out there can help me with a problem I'm having with the > Perfect Focus System on a Nikon Ti. Basically, the PFS won't work when I use > my 100X VC Plan Apochromat objective. My specimens are in various vessels, > but all of them have coverslip thickness glass and at least 3mm of > water/medium. When I focus on a specimen the PFS focus indicator light on > the front of the stand indicates that the objective lens is not in the focusable > range, even when the objective is focussed on the very surface of the > coverslip. To test this I drew crosses in fine black marker pen on the inner > coverslip surface of various chamber types. I then filled the chambers with > water and focussed on the cross using transmitted light. My results are as > follows: > > 1. The failure of the PFS seems to be limited to the 100X VC Plan Apochromat. > Lower power lenses have consistently worked when tested. > 2. If I move the stage so that the entire field is filled with the dark marker > pen, the focus indicator lights up and PFS works. This is true whether or not > the transmitted light is on or switched off. > 3. If I get the PFS to work in the manner above and then move the stage so > that I’m looking at an area free of marker pen the PFS stays on unless I turn > the transmitted light off or put an IR blocking filter in front of the transmitted > light. > 4. If I blindly (with the transmitted light off) move the stage back into an area > where there is marker pen, the focus indicator light comes back on and the > PFS will work. > 5. Additionally, if I get the PFS working using the method above and then > move the stage into a clear area – while still observing using transmitted light > – and then try to adjust the offset using the PFS Offset Controller, this causes > the PFS to fail again. > > From this I can say certain things: > > 1. Only the 100X VC Plan Apochromat is affected > 2. A dark field (or possibly one with high contrast edges) is required for the > PFS to work initially > 3. The PFS can be sustained by IR light coming from the halogen lamp, but > this isn't enough to get it going initially > 4. Adjusting the offset, which moves the dichroic mirror in the PFS nosepiece, > seems to take the system out of the focusable range > > Note that I can’t rule out a fault with the objective, but I have compared this > objective with another identical objective in other tests recently and have no > reason to believe it is faulty. > > Any ideas or suggestions would be most welcome. > > Thanks, > > Andrew. > |
|
Andrew Vaughan |
|
|
In reply to this post by Andrew Vaughan
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** First of all I'd like to thank everyone who contacted me about this on and off the forum. IR interference with the PFS was not the problem. I installed IR cut filters in both the trans and epi illumination paths and I still had the same problem. I've had two communications that seem to suggest that the particular objective lens I'm using (100X VC Plan Apochromat 1.4 NA) can cause problems for the PFS, although it is on Nikon's list of PFS-compatible objective lenses. I've managed to work around the problem by installing a 100X Plan Fluor 1.3 NA objective instead. I'm still in communication with Nikon and I'm going to try to get to the bottom of this. |
|
|
In reply to this post by Mario Faretta
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear all, we are trying to image fixed cell spheroids (>100um diameter) on a leica sp5 inverted confocal microscope. The high resolution objectives we use have short but sufficient working distance to analyse the spheres if they attached to the coverslip. However standard coatings as polilysine are not efficient in immobilizing them (even if they can initially adhere we loose them extensively during staining). We tried also particular gels as Q gel but we were not able to find a reproducible protocol to have them clearly attached to the substrate instead of being simply embedded in the gel. I know that new high NA water-immersion long working-distance objectives could provide the simplest solution but unfortunately their cost is quite limiting. thanks for all the answers Mario -- ---PLEASE Note the change in telephone number--- -- Mario Faretta Department of Experimental Oncology European Institute of Oncology c/o IFOM-IEO Campus for Oncogenomics via Adamello 16 20139 Milan Italy Phone: ++39-0294375027 email: [hidden email] http://www.ifom-ieo-campus.it |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi, Have you tried to activate your coverslip by putting it in a plasma for 30 sec to 1 min? When we do this, we find that the polylysine coating works much better than when the coverslips are only acid cleaned. Your friendly neighborhood EM lab should be able to provide a plasma for you to try. Hope it helps, Tobias >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Dear all, >we are trying to image fixed cell spheroids (>100um diameter) on a >leica sp5 inverted confocal microscope. The high resolution >objectives we use have short but sufficient working distance to >analyse the spheres if they attached to the coverslip. However >standard coatings as polilysine are not efficient in immobilizing >them (even if they can initially adhere we loose them extensively >during staining). We tried also particular gels as Q gel but we >were not able to find a reproducible protocol to have them clearly >attached to the substrate instead of being simply embedded in the >gel. >I know that new high NA water-immersion long working-distance >objectives could provide the simplest solution but unfortunately >their cost is quite limiting. >thanks for all the answers >Mario > >-- >---PLEASE Note the change in telephone number--- > >-- >Mario Faretta >Department of Experimental Oncology >European Institute of Oncology >c/o IFOM-IEO Campus for Oncogenomics >via Adamello 16 20139 Milan Italy >Phone: ++39-0294375027 >email: [hidden email] >http://www.ifom-ieo-campus.it -- _ ____ __ ____ / \ / / \ / \ \ Tobias I. Baskin / / / / \ \ \ Professor /_ / __ /__ \ \ \__ Biology Department / / / \ \ \ 611 N. Pleasant St. / / / \ \ \ University of Massachusetts / / ___ / \ \__/ \ ____ Amherst, MA, 01003 www.bio.umass.edu/biology/baskin Voice: 413 - 545 - 1533 Fax: 413 - 545 - 3243 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Tobias. what kind of plasma, like with argon or something? when you mention EM, you're thinking about the gold sputter coating device? and after, you still go through HF etching or straight to poly-L ? good to hear from you Jose Em 12-03-2012 13:32, Tobias Baskin escreveu: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi, > Have you tried to activate your coverslip by putting it in a > plasma for 30 sec to 1 min? When we do this, we find that the > polylysine coating works much better than when the coverslips are > only acid cleaned. Your friendly neighborhood EM lab should be able > to provide a plasma for you to try. > > Hope it helps, > Tobias > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear all, >> we are trying to image fixed cell spheroids (>100um diameter) on a >> leica sp5 inverted confocal microscope. The high resolution >> objectives we use have short but sufficient working distance to >> analyse the spheres if they attached to the coverslip. However >> standard coatings as polilysine are not efficient in immobilizing >> them (even if they can initially adhere we loose them extensively >> during staining). We tried also particular gels as Q gel but we >> were not able to find a reproducible protocol to have them clearly >> attached to the substrate instead of being simply embedded in the >> gel. >> I know that new high NA water-immersion long working-distance >> objectives could provide the simplest solution but unfortunately >> their cost is quite limiting. >> thanks for all the answers >> Mario >> >> -- >> ---PLEASE Note the change in telephone number--- >> >> -- >> Mario Faretta >> Department of Experimental Oncology >> European Institute of Oncology >> c/o IFOM-IEO Campus for Oncogenomics >> via Adamello 16 20139 Milan Italy >> Phone: ++39-0294375027 >> email: [hidden email] >> http://www.ifom-ieo-campus.it > -- ********************************************************** Jose' A. Feijo', Professor Catedrático ---------------------------------------------------------- Dep. Biologia Vegetal, Fac.Ciencias, Universidade Lisboa PT-1749-016 Lisboa, PORTUGAL tel. +351.21.750.00.47/00/24, fax +351.21.750.00.48 and/ e Inst.Gulbenkian Ciencia, PT-2780-156 Oeiras, PORTUGAL tel. +351.21.440.79.41/00/19, fax +351.21.440.79.70 __________________________________________________________ e.mail: [hidden email] or [hidden email] URL: http://www.igc.gulbenkian.pt/research/unit/38 or http://dbv.fc.ul.pt/users/jafeijo ********************************************************** |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi, Ah, it is true that a sputter coater used for SEM makes a plasma. This is actually fancier than what is needed (and certainly you don't want to sputter metal on your coverslip). But plasmas are also used for cleaning surfaces. Apparently they create ozone or something that attacks hydrocarbons that coat everything. It should be possible to find a 'plasma cleaner' or 'etcher' that is not a sputter coater. (Maybe on some models of sputter coaters you could make a plasma w/o coating but I think maybe not). EM labs use them activate grids and other fighing of hydroarbon menace. Materials Science labs might have one, or anyone dealing with micro (nano!) fabrication. The devices I have used are simple, they have a small chamber whre you put your coverslips or slides, and they pull a vacuum (maybe 0.1 % of air, not sure exactly but nothing too serious), and then generate a plasma in the remaining air (I guess by using RF coils but I don't know the engineering details). There is nice blue/purple glow. This makes glass extremely hydrophillic. This is useful for lots of things, including polylysine coating. Right after plasma coating coverslips, we add a drop of polylysine to the coverslip, spreading if needed, and let it air dry. Then put a generous drop of water on, let that air dry, and the coverslips are ready. We use them as soon as possible after that. The next day they are less sticky. As ever Tobias > > >Hi Tobias. what kind of plasma, like with argon >or something? when you mention EM, you're >thinking about the gold sputter coating device? >and after, you still go through HF etching or >straight to poly-L ? >good to hear from you >Jose > >Em 12-03-2012 13:32, Tobias Baskin escreveu: >> >> >>Hi, >> Have you tried to activate your coverslip by putting it in a >>plasma for 30 sec to 1 min? When we do this, we find that the >>polylysine coating works much better than when the coverslips are >>only acid cleaned. Your friendly neighborhood EM lab should be able >>to provide a plasma for you to try. >> >> Hope it helps, >> Tobias >> >>> >>>***** >>> >>>Dear all, >>>we are trying to image fixed cell spheroids (>100um diameter) on a >>>leica sp5 inverted confocal microscope. The high resolution >>>objectives we use have short but sufficient working distance to >>>analyse the spheres if they attached to the coverslip. However >>>standard coatings as polilysine are not efficient in immobilizing >>>them (even if they can initially adhere we loose them extensively >>>during staining). We tried also particular gels as Q gel but we >>>were not able to find a reproducible protocol to have them clearly >>>attached to the substrate instead of being simply embedded in the >>>gel. >>>I know that new high NA water-immersion long working-distance >>>objectives could provide the simplest solution but unfortunately >>>their cost is quite limiting. >>>thanks for all the answers >>>Mario >>> >>>-- >>>---PLEASE Note the change in telephone number--- >>> >>>-- >>>Mario Faretta >>>Department of Experimental Oncology >>>European Institute of Oncology >>>c/o IFOM-IEO Campus for Oncogenomics >>>via Adamello 16 20139 Milan Italy >>>Phone: ++39-0294375027 >>>email: [hidden email] >> > >-- > > >********************************************************** >Jose' A. Feijo', Professor Catedrático >---------------------------------------------------------- >Dep. Biologia Vegetal, Fac.Ciencias, Universidade Lisboa >PT-1749-016 Lisboa, PORTUGAL > >tel. +351.21.750.00.47/00/24, fax +351.21.750.00.48 > >and/ e > >Inst.Gulbenkian Ciencia, PT-2780-156 Oeiras, PORTUGAL > >tel. +351.21.440.79.41/00/19, fax +351.21.440.79.70 >__________________________________________________________ >e.mail: [hidden email] or [hidden email] > > >********************************************************** -- _ ____ __ ____ / \ / / \ / \ \ Tobias I. Baskin / / / / \ \ \ Professor /_ / __ /__ \ \ \__ Biology Department / / / \ \ \ 611 N. Pleasant St. / / / \ \ \ University of Massachusetts / / ___ / \ \__/ \ ____ Amherst, MA, 01003 www.bio.umass.edu/biology/baskin Voice: 413 - 545 - 1533 Fax: 413 - 545 - 3243 |
|
Boswell, Carl A - (cboswell) |
|
|
In reply to this post by Mario Faretta
Hi Mario,
Celltak is an excellent biological adhesive and would be particularly appropriate for your fixed preps. It is the glue that marine mussels use to attach to rocks in the intertidal (read that "high wave action") zone. There can be some problems with live cells as the adhesion is so strong that normal physiology can be disturbed. http://www.bdbiosciences.com/external_files/dl/doc/manuals/live/web_enabled/354240_pug.pdf C Carl A. Boswell 520-742-6131 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Mario Faretta Sent: Monday, March 12, 2012 5:42 AM To: [hidden email] Subject: mammospheres attachment ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear all, we are trying to image fixed cell spheroids (>100um diameter) on a leica sp5 inverted confocal microscope. The high resolution objectives we use have short but sufficient working distance to analyse the spheres if they attached to the coverslip. However standard coatings as polilysine are not efficient in immobilizing them (even if they can initially adhere we loose them extensively during staining). We tried also particular gels as Q gel but we were not able to find a reproducible protocol to have them clearly attached to the substrate instead of being simply embedded in the gel. I know that new high NA water-immersion long working-distance objectives could provide the simplest solution but unfortunately their cost is quite limiting. thanks for all the answers Mario -- ---PLEASE Note the change in telephone number--- -- Mario Faretta Department of Experimental Oncology European Institute of Oncology c/o IFOM-IEO Campus for Oncogenomics via Adamello 16 20139 Milan Italy Phone: ++39-0294375027 email: [hidden email] http://www.ifom-ieo-campus.it |
| Free forum by Nabble | Edit this page |