Does anyone know what the light transmission efficiency of the Nuance multispectral imaging system from Perkin Elmer is?
I've been trying to find a % transmission curve for the liquid crystal filter but have not been able to.

Sincerely,

Claire
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Nuance Efficiency Curve

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Liquid crystal filters usually only work with polarized light, so the first
part of the filter system is usually a linear polarizer. This will
immediately take out half your light, assuming unpolarized light from your
sample. The LC also has a reduced transmission on top of that so overall
the system efficiencies will be mediocre at best.

Craig
On Jun 21, 2016 9:24 AM, "Claire Brown, Dr." <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Does anyone know what the light transmission efficiency of the Nuance
> multispectral imaging system from Perkin Elmer is?
> I've been trying to find a % transmission curve for the liquid crystal
> filter but have not been able to.
>
> Sincerely,
>
> Claire
>
George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Nuance Efficiency Curve

In reply to this post by Claire Brown
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Claire,

Figure 1, top panel, is the relevant curve for the Nuance and Vectra
LCTF (note this is for properly polarized light, divide Y axis byt 2 for
non-polarized light).

http://www.spectralcameras.com/files/VariSpec_Brochure.pdf

Craig noted in his reply: "Liquid crystal filters usually only work with
polarized light, so the first part of the filter system is usually a
linear polarizer. This will immediately take out half your light,
assuming unpolarized light from your sample. The LC also has a reduced
transmission on top of that so overall the system efficiencies will be
mediocre at best"

Richard Levenson, Joe Beechem, and I wrote a review (open access) on
spectral imaging that discusses various options (a fast alternative,
AOTF, is also polarization dependent):

http://www.hindawi.com/journals/acp/2012/904828/abs/

and finishes with my favorite "SKY400" chromosome paints image (in
answer to another thread: Applied Spectral Imaging and several other
companies sell each and all human chromosome paint probe sets). As for
section 14 "going for 100-plex" ... hopefully Garry Nolan will get his
"MIBI-TOF" paper out (probably only 40plex; near single ion sensitivity,
which is fine since each antibody has ~180 mass tags).

enjoy,

George
p.s. LCTFing non-polarized light, PerkinElmer could use a polarizing
beamsplitter to split the V and H pol, put those through separate LCTFs
(of proper orientations), then recombine onto one camera ("Fluorospec"),
or send to two cameras. The latter would be a lot of fun for "spectral
fluorescence anisotropy".


On 6/20/2016 1:33 PM, Claire Brown, Dr. wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Does anyone know what the light transmission efficiency of the Nuance multispectral imaging system from Perkin Elmer is?
> I've been trying to find a % transmission curve for the liquid crystal filter but have not been able to.
>
> Sincerely,
>
> Claire
>


--



George McNamara, Ph.D.
Houston, TX 77054
[hidden email]
713-239-0365 home
305-764-2081 cell
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/   Tattletales and T-Bow
Sripad Ram-2 Sripad Ram-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Nuance Efficiency Curve

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hello,
The Varispec LCTF is only half the part of the nuance/Vectra camera. You
still need to factor in the QE of the CCD chip. I vaguely recall seeing a
PPT presentation that the Nuance uses a Sony ICX285 (or similar) chip. So
max QE will be around ~70% at 500 nm.

HTH.

Sripad


On Tue, Jun 21, 2016 at 11:13 AM, George McNamara <[hidden email]
> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Claire,
>
> Figure 1, top panel, is the relevant curve for the Nuance and Vectra LCTF
> (note this is for properly polarized light, divide Y axis byt 2 for
> non-polarized light).
>
> http://www.spectralcameras.com/files/VariSpec_Brochure.pdf
>
> Craig noted in his reply: "Liquid crystal filters usually only work with
> polarized light, so the first part of the filter system is usually a linear
> polarizer. This will immediately take out half your light, assuming
> unpolarized light from your sample. The LC also has a reduced transmission
> on top of that so overall the system efficiencies will be mediocre at best"
>
> Richard Levenson, Joe Beechem, and I wrote a review (open access) on
> spectral imaging that discusses various options (a fast alternative, AOTF,
> is also polarization dependent):
>
> http://www.hindawi.com/journals/acp/2012/904828/abs/
>
> and finishes with my favorite "SKY400" chromosome paints image (in answer
> to another thread: Applied Spectral Imaging and several other companies
> sell each and all human chromosome paint probe sets). As for section 14
> "going for 100-plex" ... hopefully Garry Nolan will get his "MIBI-TOF"
> paper out (probably only 40plex; near single ion sensitivity, which is fine
> since each antibody has ~180 mass tags).
>
> enjoy,
>
> George
> p.s. LCTFing non-polarized light, PerkinElmer could use a polarizing
> beamsplitter to split the V and H pol, put those through separate LCTFs (of
> proper orientations), then recombine onto one camera ("Fluorospec"), or
> send to two cameras. The latter would be a lot of fun for "spectral
> fluorescence anisotropy".
>
>
>
> On 6/20/2016 1:33 PM, Claire Brown, Dr. wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> Post images on http://www.imgur.com and include the link in your posting.
>> *****
>>
>> Does anyone know what the light transmission efficiency of the Nuance
>> multispectral imaging system from Perkin Elmer is?
>> I've been trying to find a % transmission curve for the liquid crystal
>> filter but have not been able to.
>>
>> Sincerely,
>>
>> Claire
>>
>>
>
> --
>
>
>
> George McNamara, Ph.D.
> Houston, TX 77054
> [hidden email]
> 713-239-0365 home
> 305-764-2081 cell
> https://www.linkedin.com/in/georgemcnamara
> https://works.bepress.com/gmcnamara/75/   Tattletales and T-Bow
>
Loralei Dewe-3 Loralei Dewe-3
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Nuance Efficiency Curve

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hi Claire,

It's been a long time....I hope all is well;-)

Another thing besides the QE with the Nuance camera is that it is only a 10
bit camera which severely limits the dynamic range. I have a lab at
Stanford that I installed a Flash 4.0 V2 in with a dual adapter and both
the cameras are running in Metamorph.  Besides the FOV and dynamic range
they work together well.

I have a great contact in PE that was around when CRi developed the Nuance
and would be happy to share if you'd be interested?

Loralei Dewe
Chrysalis Innovations
On Jun 21, 2016 1:34 PM, "S Ram" <[hidden email]> wrote:

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hello,
The Varispec LCTF is only half the part of the nuance/Vectra camera. You
still need to factor in the QE of the CCD chip. I vaguely recall seeing a
PPT presentation that the Nuance uses a Sony ICX285 (or similar) chip. So
max QE will be around ~70% at 500 nm.

HTH.

Sripad


On Tue, Jun 21, 2016 at 11:13 AM, George McNamara <[hidden email]
> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
>
> Hi Claire,
>
> Figure 1, top panel, is the relevant curve for the Nuance and Vectra LCTF
> (note this is for properly polarized light, divide Y axis byt 2 for
> non-polarized light).
>
> http://www.spectralcameras.com/files/VariSpec_Brochure.pdf
>
> Craig noted in his reply: "Liquid crystal filters usually only work with
> polarized light, so the first part of the filter system is usually a
linear
> polarizer. This will immediately take out half your light, assuming
> unpolarized light from your sample. The LC also has a reduced transmission
> on top of that so overall the system efficiencies will be mediocre at
best"

>
> Richard Levenson, Joe Beechem, and I wrote a review (open access) on
> spectral imaging that discusses various options (a fast alternative, AOTF,
> is also polarization dependent):
>
> http://www.hindawi.com/journals/acp/2012/904828/abs/
>
> and finishes with my favorite "SKY400" chromosome paints image (in answer
> to another thread: Applied Spectral Imaging and several other companies
> sell each and all human chromosome paint probe sets). As for section 14
> "going for 100-plex" ... hopefully Garry Nolan will get his "MIBI-TOF"
> paper out (probably only 40plex; near single ion sensitivity, which is
fine
> since each antibody has ~180 mass tags).
>
> enjoy,
>
> George
> p.s. LCTFing non-polarized light, PerkinElmer could use a polarizing
> beamsplitter to split the V and H pol, put those through separate LCTFs
(of

> proper orientations), then recombine onto one camera ("Fluorospec"), or
> send to two cameras. The latter would be a lot of fun for "spectral
> fluorescence anisotropy".
>
>
>
> On 6/20/2016 1:33 PM, Claire Brown, Dr. wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> Post images on http://www.imgur.com and include the link in your posting.
>> *****
>>
>> Does anyone know what the light transmission efficiency of the Nuance
>> multispectral imaging system from Perkin Elmer is?
>> I've been trying to find a % transmission curve for the liquid crystal
>> filter but have not been able to.
>>
>> Sincerely,
>>
>> Claire
>>
>>
>
> --
>
>
>
> George McNamara, Ph.D.
> Houston, TX 77054
> [hidden email]
> 713-239-0365 home
> 305-764-2081 cell
> https://www.linkedin.com/in/georgemcnamara
> https://works.bepress.com/gmcnamara/75/   Tattletales and T-Bow
>