Objective - maximum laser power

classic Classic list List threaded Threaded
7 messages Options
Steffen Steinert Steffen Steinert
Reply | Threaded
Open this post in threaded view
|

Objective - maximum laser power

Hello,

 

I was wondering whether someone tried to focus the excitation laser beam to the backfocal plane of the objective (60xoil) for widefield illumination. If so what was the maximum power used without affecting the functionality of the objective?

I´ve a max power of 3W cw @ 532nm with an estimated focus at the backfocal plane of ~1mm but I´m not eager to destroy the objective. Any ideas/advices/experiences how far one can go regarding the laser excitation??

 

Many thanks in advance,

 

Steffen  

 

Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
|

Re: Objective - maximum laser power

Is this CW or pulsed laser power?  Pulsed lasers can be more damaging
for an equivalent CW average power.

Craig


On Thu, Jan 28, 2010 at 2:35 AM, Steffen Steinert
<[hidden email]> wrote:

> Hello,
>
>
>
> I was wondering whether someone tried to focus the excitation laser beam to
> the backfocal plane of the objective (60xoil) for widefield illumination. If
> so what was the maximum power used without affecting the functionality of
> the objective?
>
> I´ve a max power of 3W cw @ 532nm with an estimated focus at the backfocal
> plane of ~1mm but I´m not eager to destroy the objective. Any
> ideas/advices/experiences how far one can go regarding the laser
> excitation??
>
>
>
> Many thanks in advance,
>
>
>
> Steffen
>
>
Steffen Steinert Steffen Steinert
Reply | Threaded
Open this post in threaded view
|

AW: Objective - maximum laser power

Hi Craig,

it´s cw-laser! At a later stage I´m going to run some complicated sequences
by shutting off the laser with an AOM, but that´s a different story. At the
moment I´d just like to do experiments at cw without damaging the objective.
I tried 100mW without any problem (so far), but I´d actually need some more
excitation power...

Thanks for any input!

Steffen

-----Ursprüngliche Nachricht-----
Von: Confocal Microscopy List [mailto:[hidden email]] Im
Auftrag von Craig Brideau
Gesendet: Donnerstag, 28. Januar 2010 18:50
An: [hidden email]
Betreff: Re: Objective - maximum laser power

Is this CW or pulsed laser power?  Pulsed lasers can be more damaging
for an equivalent CW average power.

Craig


On Thu, Jan 28, 2010 at 2:35 AM, Steffen Steinert
<[hidden email]> wrote:
> Hello,
>
>
>
> I was wondering whether someone tried to focus the excitation laser beam
to
> the backfocal plane of the objective (60xoil) for widefield illumination.
If

> so what was the maximum power used without affecting the functionality of
> the objective?
>
> I´ve a max power of 3W cw @ 532nm with an estimated focus at the backfocal
> plane of ~1mm but I´m not eager to destroy the objective. Any
> ideas/advices/experiences how far one can go regarding the laser
> excitation??
>
>
>
> Many thanks in advance,
>
>
>
> Steffen
>
>
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
|

Re: Objective - maximum laser power

In reply to this post by Craig Brideau
Well, I put >400mW of 950nm at 100fs through my microscope with no
trouble, so you should have no problems increasing your power a bit.
I could probably even risk a bit more power if I had to although I'd
prefer to keep things under 400mW.  With CW visible you should be able
to do around that.  To be careful though you should probably check
with the manufacturer if possible.  Your mileage may vary, do at own
risk, buyer beware, etc. @:-)  Check to see if your dichroics and
other filtering mechanisms can handle it as they will probably be the
most fragile components.

Craig


On Thu, Jan 28, 2010 at 10:56 AM, Steffen Steinert
<[hidden email]> wrote:

> Hi Craig,
>
> it´s cw-laser! At a later stage I´m going to run some complicated sequences
> by shutting off the laser with an AOM, but that´s a different story. At the
> moment I´d just like to do experiments at cw without damaging the objective.
> I tried 100mW without any problem (so far), but I´d actually need some more
> excitation power...
>
> Thanks for any input!
>
> Steffen
>
> -----Ursprüngliche Nachricht-----
> Von: Confocal Microscopy List [mailto:[hidden email]] Im
> Auftrag von Craig Brideau
> Gesendet: Donnerstag, 28. Januar 2010 18:50
> An: [hidden email]
> Betreff: Re: Objective - maximum laser power
>
> Is this CW or pulsed laser power?  Pulsed lasers can be more damaging
> for an equivalent CW average power.
>
> Craig
>
>
> On Thu, Jan 28, 2010 at 2:35 AM, Steffen Steinert
> <[hidden email]> wrote:
>> Hello,
>>
>>
>>
>> I was wondering whether someone tried to focus the excitation laser beam
> to
>> the backfocal plane of the objective (60xoil) for widefield illumination.
> If
>> so what was the maximum power used without affecting the functionality of
>> the objective?
>>
>> I´ve a max power of 3W cw @ 532nm with an estimated focus at the backfocal
>> plane of ~1mm but I´m not eager to destroy the objective. Any
>> ideas/advices/experiences how far one can go regarding the laser
>> excitation??
>>
>>
>>
>> Many thanks in advance,
>>
>>
>>
>> Steffen
>>
>>
>
Jan Grawe Jan Grawe
Reply | Threaded
Open this post in threaded view
|

Re: Objective - maximum laser power

In reply to this post by Steffen Steinert

Hello Steffen,

 

We have done 1.2W of 532nm CW through a Zeiss Fluar 40x oil for hours without any appreciable detoriation. However, we are also a bit apprehensive about the long-term effects on resins etc in the objective.

 

/Janne

 

Jan Grawé

Cell Analysis Core Facility

Rudbecklaboratoriet/C5

Dag Hammarskjölds väg 20

SE-75185 Uppsala

SWEDEN

Phone: +(0)18-4714656

Cell:+(0)70-2577874

[hidden email]

http://www.rudbeck.uu.se/cellanalys

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Steffen Steinert
Sent: den 28 januari 2010 10:35
To: [hidden email]
Subject: Objective - maximum laser power

 

Hello,

 

I was wondering whether someone tried to focus the excitation laser beam to the backfocal plane of the objective (60xoil) for widefield illumination. If so what was the maximum power used without affecting the functionality of the objective?

I´ve a max power of 3W cw @ 532nm with an estimated focus at the backfocal plane of ~1mm but I´m not eager to destroy the objective. Any ideas/advices/experiences how far one can go regarding the laser excitation??

 

Many thanks in advance,

 

Steffen  

 

Armstrong, Brian Armstrong, Brian
Reply | Threaded
Open this post in threaded view
|

Re: Objective - maximum laser power

Hello, I am just curious, why do you do this?

 

 

Brian D Armstrong PhD

Light Microscopy Core Manager

Beckman Research Institute

City of Hope

Dept of Neuroscience

1450 E Duarte Rd

Duarte, CA 91010

626-256-4673 x62872

http://www.cityofhope.org/research/support/Light-Microscopy-Digital-Imaging/Pages/default.aspx


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Jan Grawe
Sent: Friday, January 29, 2010 12:21 AM
To: [hidden email]
Subject: Re: Objective - maximum laser power

 

Hello Steffen,

 

We have done 1.2W of 532nm CW through a Zeiss Fluar 40x oil for hours without any appreciable detoriation. However, we are also a bit apprehensive about the long-term effects on resins etc in the objective.

 

/Janne

 

Jan Grawé

Cell Analysis Core Facility

Rudbecklaboratoriet/C5

Dag Hammarskjölds väg 20

SE-75185 Uppsala

SWEDEN

Phone: +(0)18-4714656

Cell:+(0)70-2577874

[hidden email]

http://www.rudbeck.uu.se/cellanalys

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Steffen Steinert
Sent: den 28 januari 2010 10:35
To: [hidden email]
Subject: Objective - maximum laser power

 

Hello,

 

I was wondering whether someone tried to focus the excitation laser beam to the backfocal plane of the objective (60xoil) for widefield illumination. If so what was the maximum power used without affecting the functionality of the objective?

I´ve a max power of 3W cw @ 532nm with an estimated focus at the backfocal plane of ~1mm but I´m not eager to destroy the objective. Any ideas/advices/experiences how far one can go regarding the laser excitation??

 

Many thanks in advance,

 

Steffen  

 


---------------------------------------------------------------------
SECURITY/CONFIDENTIALITY WARNING:
This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender.
---------------------------------------------------------------------

Steffen Steinert Steffen Steinert
Reply | Threaded
Open this post in threaded view
|

AW: Objective - maximum laser power

Hello, I am just curious, why do you do this?

- Well, I´m actually using a non-bleachable fluorescent probe, the so called nitrogen-vacancy center in diamond.

Besides other interesting features one can optically readout magnetic fields. In order to saturate all defect centers in widefield illumination, it needs a lot of excitation power.

 

Thanks to all who responded, much appreciated!

 

Best regards,

 

Steffen

 

Von: Confocal Microscopy List [mailto:[hidden email]] Im Auftrag von Armstrong, Brian
Gesendet: Freitag, 29. Januar 2010 18:04
An: [hidden email]
Betreff: Re: Objective - maximum laser power

 

 

Hello, I am just curious, why do you do this?

 

Brian D Armstrong PhD

Light Microscopy Core Manager

Beckman Research Institute

City of Hope

Dept of Neuroscience

1450 E Duarte Rd

Duarte, CA 91010

626-256-4673 x62872

http://www.cityofhope.org/research/support/Light-Microscopy-Digital-Imaging/Pages/default.aspx


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Jan Grawe
Sent: Friday, January 29, 2010 12:21 AM
To: [hidden email]
Subject: Re: Objective - maximum laser power

 

Hello Steffen,

 

We have done 1.2W of 532nm CW through a Zeiss Fluar 40x oil for hours without any appreciable detoriation. However, we are also a bit apprehensive about the long-term effects on resins etc in the objective.

 

/Janne

 

Jan Grawé

Cell Analysis Core Facility

Rudbecklaboratoriet/C5

Dag Hammarskjölds väg 20

SE-75185 Uppsala

SWEDEN

Phone: +(0)18-4714656

Cell:+(0)70-2577874

[hidden email]

http://www.rudbeck.uu.se/cellanalys

 

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Steffen Steinert
Sent: den 28 januari 2010 10:35
To: [hidden email]
Subject: Objective - maximum laser power

 

Hello,

 

I was wondering whether someone tried to focus the excitation laser beam to the backfocal plane of the objective (60xoil) for widefield illumination. If so what was the maximum power used without affecting the functionality of the objective?

I´ve a max power of 3W cw @ 532nm with an estimated focus at the backfocal plane of ~1mm but I´m not eager to destroy the objective. Any ideas/advices/experiences how far one can go regarding the laser excitation??

 

Many thanks in advance,

 

Steffen  

 


---------------------------------------------------------------------
SECURITY/CONFIDENTIALITY WARNING:
This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender.
---------------------------------------------------------------------