![]() ![]() |
Gary Laevsky |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi All, Trying not to be cynical here, but, can anybody please explain to me if there is a way for a drop of oil to end up on the center of the back focal plane of an objective (oil immersion objective) on an inverted scope. No smear from the side. Thanks. -- Best, Gary Laevsky, Ph.D. Director, Confocal Imaging Facility Nikon Center of Excellence Dept. of Molecular Biology Washington Rd. Princeton University Princeton, New Jersey, 08544-1014 (O) 609 258 5432 (C) 508 507 1310 |
![]() ![]() |
John Oreopoulos |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I've never experienced it myself, but if it's a spring-loaded objective barrel, oil can potentially leak through the gap between the spring-loaded section and the outside barrel and make it's way down through the objective - maybe. This is why I wipe oil off objectives as soon as I'm done using them. Hard to imagine any other way oil could get there, unless maybe oil dropped onto the nosepiece at the base of the objective? Or maybe someone removed the objective to clean it and accidentally set it down on a surface with oil? In any case, it should be fine to clean it off it it's only on the outside surface of the bottom lens of the objective. Sending good vibes your way, John Oreopoulos On 2016-04-04, at 5:48 AM, Gary Laevsky wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi All, > > Trying not to be cynical here, but, can anybody please explain to me if > there is a way for a drop of oil to end up on the center of the back focal > plane of an objective (oil immersion objective) on an inverted scope. No > smear from the side. > > Thanks. > > -- > Best, > > Gary Laevsky, Ph.D. > Director, Confocal Imaging Facility > Nikon Center of Excellence > Dept. of Molecular Biology > Washington Rd. > Princeton University > Princeton, New Jersey, 08544-1014 > (O) 609 258 5432 > (C) 508 507 1310 |
![]() |
In reply to this post by Gary Laevsky
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Oil can sometimes leak inside objectives, especially on those with anti-shock... - not sure what it's called, when the central part can move up and down. We once had to throw away one quite expensive piece. It helps to wipe objectives after use. Mike ________________________________________ From: Confocal Microscopy List <[hidden email]> on behalf of Gary Laevsky <[hidden email]> Sent: Monday, April 4, 2016 5:48 AM To: [hidden email] Subject: Oil on back focal plane of objective ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi All, Trying not to be cynical here, but, can anybody please explain to me if there is a way for a drop of oil to end up on the center of the back focal plane of an objective (oil immersion objective) on an inverted scope. No smear from the side. Thanks. -- Best, Gary Laevsky, Ph.D. Director, Confocal Imaging Facility Nikon Center of Excellence Dept. of Molecular Biology Washington Rd. Princeton University Princeton, New Jersey, 08544-1014 (O) 609 258 5432 (C) 508 507 1310 |
![]() ![]() |
Oshel, Philip Eugene |
![]() |
In reply to this post by Gary Laevsky
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I've noticed the same thing. Discarding the grumpy usual, on the basis of not having "people like that" using the confocals, the only thing I can think of is minute pinholes in the mounting cement allowing the oil to soak through - by gravity if on on inverted microscope, or capillary action if on an upright. Phil On 04/04/2016 05:48 , Gary Laevsky wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi All, > > Trying not to be cynical here, but, can anybody please explain to me if > there is a way for a drop of oil to end up on the center of the back focal > plane of an objective (oil immersion objective) on an inverted scope. No > smear from the side. > > Thanks. > -- Philip Oshel Microscopy Facility Supervisor Biology Department 024C Brooks Hall Central Michigan University Mt. Pleasant, MI 48859 (989) 774-3576 |
![]() |
This would seem to be physically impossible on a BF objective - the oil would have to levitate! So I suppose we are talking about a phase objective - but one really shouldn't use a phase objective for confocal imaging. Having said that I have to agree with the interpretation - if you must use an inverted microscope it is crucial not just to clean the oil off after use, but to then remove the objective and store it tip down until it's next needed. That way anything that has leaked in gets a chance to leak out again. What's more, I think I (or maybe Ellie Kable) posted the same message some 15 years ago .....
Guy Guy Cox, Honorary Associate Professor School of Medical Sciences Australian Centre for Microscopy and Microanalysis, Madsen, F09, University of Sydney, NSW 2006 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Philip Oshel Sent: Monday, 4 April 2016 10:27 PM To: [hidden email] Subject: Re: Oil on back focal plane of objective ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I've noticed the same thing. Discarding the grumpy usual, on the basis of not having "people like that" using the confocals, the only thing I can think of is minute pinholes in the mounting cement allowing the oil to soak through - by gravity if on on inverted microscope, or capillary action if on an upright. Phil On 04/04/2016 05:48 , Gary Laevsky wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi All, > > Trying not to be cynical here, but, can anybody please explain to me > if there is a way for a drop of oil to end up on the center of the > back focal plane of an objective (oil immersion objective) on an > inverted scope. No smear from the side. > > Thanks. > -- Philip Oshel Microscopy Facility Supervisor Biology Department 024C Brooks Hall Central Michigan University Mt. Pleasant, MI 48859 (989) 774-3576 |
![]() ![]() |
Gary Laevsky |
![]() |
In reply to this post by Gary Laevsky
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi All, To be clear, if oil was leaking down the barrel of the objective, and ended up on the center of the back focal plan, there would be a trail from the edge, right? This was a solitary drop on the center of the back. I'm trying to find someone that will correct me from thinking this was a malicious act. Best, gary On Mon, Apr 4, 2016 at 5:48 AM, Gary Laevsky <[hidden email]> wrote: > Hi All, > > Trying not to be cynical here, but, can anybody please explain to me if > there is a way for a drop of oil to end up on the center of the back focal > plane of an objective (oil immersion objective) on an inverted scope. No > smear from the side. > > Thanks. > > -- > Best, > > Gary Laevsky, Ph.D. > Director, Confocal Imaging Facility > Nikon Center of Excellence > Dept. of Molecular Biology > Washington Rd. > Princeton University > Princeton, New Jersey, 08544-1014 > (O) 609 258 5432 > (C) 508 507 1310 > -- Best, Gary Laevsky, Ph.D. Director, Confocal Imaging Facility Nikon Center of Excellence Dept. of Molecular Biology Washington Rd. Princeton University Princeton, New Jersey, 08544-1014 (O) 609 258 5432 (C) 508 507 1310 |
![]() ![]() |
Jordan Becker |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hey, What about inexperienced or rushed hands? Oil on a finger accidentally grazing the bottom of the rear lens when placing/installing or removing the objective? Unless it's clearly a drop...is it a smudge/print shape? On Mon, Apr 4, 2016 at 9:39 AM, Gary Laevsky <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi All, > > To be clear, if oil was leaking down the barrel of the objective, and ended > up on the center of the back focal plan, there would be a trail from the > edge, right? > > This was a solitary drop on the center of the back. > > I'm trying to find someone that will correct me from thinking this was a > malicious act. > > Best, > > gary > > On Mon, Apr 4, 2016 at 5:48 AM, Gary Laevsky <[hidden email]> > wrote: > > > Hi All, > > > > Trying not to be cynical here, but, can anybody please explain to me if > > there is a way for a drop of oil to end up on the center of the back > focal > > plane of an objective (oil immersion objective) on an inverted scope. No > > smear from the side. > > > > Thanks. > > > > -- > > Best, > > > > Gary Laevsky, Ph.D. > > Director, Confocal Imaging Facility > > Nikon Center of Excellence > > Dept. of Molecular Biology > > Washington Rd. > > Princeton University > > Princeton, New Jersey, 08544-1014 > > (O) 609 258 5432 > > (C) 508 507 1310 > > > > > > -- > Best, > > Gary Laevsky, Ph.D. > Director, Confocal Imaging Facility > Nikon Center of Excellence > Dept. of Molecular Biology > Washington Rd. > Princeton University > Princeton, New Jersey, 08544-1014 > (O) 609 258 5432 > (C) 508 507 1310 > |
![]() ![]() |
Zdenek Svindrych-2 |
![]() |
In reply to this post by Gary Laevsky
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi. To me it seems very unlikely someone would do that on purpose. It's more likely that some greenhorn user misunderstood the sentence "don't forget to apply oil to the oil immersion objective" or something like that in your manual (I'm sure you don't specifically note that it's not necessary to unscrew the objective from the turret to do that :-). But seriously, if the oil got inside the objective through the front lens seal and found its way to the last lens, the lens is probably full of oil and can't be fixed. You may try to clean it and test whether it performs well, or ask the vendor whether they can check it for you (and ask how much it would cost!). If you have service contract, they might do it for you for free, you might even be covered by some sort of warranty. I doubt oil could get there from the lens threads through capillary action, and oil is not superfluid... Good Luck! zdenek ---------- Původní zpráva ---------- Od: Gary Laevsky <[hidden email]> Komu: [hidden email] Datum: 4. 4. 2016 10:39:50 Předmět: Re: Oil on back focal plane of objective "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi All, To be clear, if oil was leaking down the barrel of the objective, and ended up on the center of the back focal plan, there would be a trail from the edge, right? This was a solitary drop on the center of the back. I'm trying to find someone that will correct me from thinking this was a malicious act. Best, gary On Mon, Apr 4, 2016 at 5:48 AM, Gary Laevsky <[hidden email]> wrote: > Hi All, > > Trying not to be cynical here, but, can anybody please explain to me if > there is a way for a drop of oil to end up on the center of the back focal > plane of an objective (oil immersion objective) on an inverted scope. No > smear from the side. > > Thanks. > > -- > Best, > > Gary Laevsky, Ph.D. > Director, Confocal Imaging Facility > Nikon Center of Excellence > Dept. of Molecular Biology > Washington Rd. > Princeton University > Princeton, New Jersey, 08544-1014 > (O) 609 258 5432 > (C) 508 507 1310 > -- Best, Gary Laevsky, Ph.D. Director, Confocal Imaging Facility Nikon Center of Excellence Dept. of Molecular Biology Washington Rd. Princeton University Princeton, New Jersey, 08544-1014 (O) 609 258 5432 (C) 508 507 1310" |
![]() ![]() |
David Claypool |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** To All, This is a somewhat common problem with oil immersion objectives on inverted microscopes and is not limited to any one brand or manufacturer. If excess oil is not removed after using the objective m(a simple wipe with lens tissue will suffice), gravity will eventually cause the oil to run down the spring loaded front lens barrel into the objective. Over time enough oil will build up in the objective, so that it can find its way between the inner lens elements. The other scenario is if the front lens of the objective has been crashed into a specimen holder or the underside of a x-y stage. This can cause the seal around the front lens to develop a microcrack and the immersion oil will find its way into the objective. Either of the above damage is usually repairable by the manufacturer, but not under warranty. I have seen people use a piece of a pipe cleaner wrapped around the objective to catch the excess oil, but you need to remember to change it regularly. David J. Claypool Digital Imaging Product Manager Micro Video Instruments Office: 800-875-2041 x5221 Cell: 603-809-5342 [hidden email] > On Apr 4, 2016, at 11:16 AM, Zdenek Svindrych <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > > Hi. > To me it seems very unlikely someone would do that on purpose. It's more > likely that some greenhorn user misunderstood the sentence "don't forget to > apply oil to the oil immersion objective" or something like that in your > manual (I'm sure you don't specifically note that it's not necessary to > unscrew the objective from the turret to do that :-). > But seriously, if the oil got inside the objective through the front lens > seal and found its way to the last lens, the lens is probably full of oil > and can't be fixed. You may try to clean it and test whether it performs > well, or ask the vendor whether they can check it for you (and ask how much > it would cost!). If you have service contract, they might do it for you for > free, you might even be covered by some sort of warranty. > > I doubt oil could get there from the lens threads through capillary action, > and oil is not superfluid... > > Good Luck! > > zdenek > > > ---------- Původní zpráva ---------- > Od: Gary Laevsky <[hidden email]> > Komu: [hidden email] > Datum: 4. 4. 2016 10:39:50 > Předmět: Re: Oil on back focal plane of objective > > "***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi All, > > To be clear, if oil was leaking down the barrel of the objective, and ended > up on the center of the back focal plan, there would be a trail from the > edge, right? > > This was a solitary drop on the center of the back. > > I'm trying to find someone that will correct me from thinking this was a > malicious act. > > Best, > > gary > > On Mon, Apr 4, 2016 at 5:48 AM, Gary Laevsky <[hidden email]> > wrote: > >> Hi All, >> >> Trying not to be cynical here, but, can anybody please explain to me if >> there is a way for a drop of oil to end up on the center of the back focal >> plane of an objective (oil immersion objective) on an inverted scope. No >> smear from the side. >> >> Thanks. >> >> -- >> Best, >> >> Gary Laevsky, Ph.D. >> Director, Confocal Imaging Facility >> Nikon Center of Excellence >> Dept. of Molecular Biology >> Washington Rd. >> Princeton University >> Princeton, New Jersey, 08544-1014 >> (O) 609 258 5432 >> (C) 508 507 1310 > > > > -- > Best, > > Gary Laevsky, Ph.D. > Director, Confocal Imaging Facility > Nikon Center of Excellence > Dept. of Molecular Biology > Washington Rd. > Princeton University > Princeton, New Jersey, 08544-1014 > (O) 609 258 5432 > (C) 508 507 1310" |
![]() ![]() |
Gary Laevsky |
![]() |
In reply to this post by Zdenek Svindrych-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Ok, I'm going with oil leaking down the side of the barrel, and the trail disappeared (I would have xpected to see the trail). Thanks for the clarity, and restoration. Best, Gary On Mon, Apr 4, 2016 at 11:15 AM, Zdenek Svindrych <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > > Hi. > To me it seems very unlikely someone would do that on purpose. It's more > likely that some greenhorn user misunderstood the sentence "don't forget to > apply oil to the oil immersion objective" or something like that in your > manual (I'm sure you don't specifically note that it's not necessary to > unscrew the objective from the turret to do that :-). > But seriously, if the oil got inside the objective through the front lens > seal and found its way to the last lens, the lens is probably full of oil > and can't be fixed. You may try to clean it and test whether it performs > well, or ask the vendor whether they can check it for you (and ask how much > it would cost!). If you have service contract, they might do it for you for > free, you might even be covered by some sort of warranty. > > I doubt oil could get there from the lens threads through capillary action, > and oil is not superfluid... > > Good Luck! > > zdenek > > > ---------- Původní zpráva ---------- > Od: Gary Laevsky <[hidden email]> > Komu: [hidden email] > Datum: 4. 4. 2016 10:39:50 > Předmět: Re: Oil on back focal plane of objective > > "***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi All, > > To be clear, if oil was leaking down the barrel of the objective, and ended > up on the center of the back focal plan, there would be a trail from the > edge, right? > > This was a solitary drop on the center of the back. > > I'm trying to find someone that will correct me from thinking this was a > malicious act. > > Best, > > gary > > On Mon, Apr 4, 2016 at 5:48 AM, Gary Laevsky <[hidden email]> > wrote: > > > Hi All, > > > > Trying not to be cynical here, but, can anybody please explain to me if > > there is a way for a drop of oil to end up on the center of the back > focal > > plane of an objective (oil immersion objective) on an inverted scope. No > > smear from the side. > > > > Thanks. > > > > -- > > Best, > > > > Gary Laevsky, Ph.D. > > Director, Confocal Imaging Facility > > Nikon Center of Excellence > > Dept. of Molecular Biology > > Washington Rd. > > Princeton University > > Princeton, New Jersey, 08544-1014 > > (O) 609 258 5432 > > (C) 508 507 1310 > > > > > > -- > Best, > > Gary Laevsky, Ph.D. > Director, Confocal Imaging Facility > Nikon Center of Excellence > Dept. of Molecular Biology > Washington Rd. > Princeton University > Princeton, New Jersey, 08544-1014 > (O) 609 258 5432 > (C) 508 507 1310" > -- Best, Gary Laevsky, Ph.D. Director, Confocal Imaging Facility Nikon Center of Excellence Dept. of Molecular Biology Washington Rd. Princeton University Princeton, New Jersey, 08544-1014 (O) 609 258 5432 (C) 508 507 1310 |
![]() ![]() |
Craig Brideau |
![]() |
In reply to this post by David Claypool
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I still stand by my recommendation of cloth hair scrunchies on the objectives. It catches any oil drips running down the side of the objective on inverted scopes. It won't save you from oil leaking into the spring-loaded tips but it at least keeps it from soaking into the optics under the objective. You can even use them for color coding! http://previews.123rf.com/images/belchonock/belchonock1108/belchonock110800172/10220322-hair-scrunchies-on-brown-background-Stock-Photo.jpg Craig Brideau On Mon, Apr 4, 2016 at 9:31 AM, David Claypool <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > To All, > > This is a somewhat common problem with oil immersion objectives on > inverted microscopes and is not limited to any one brand or manufacturer. > > If excess oil is not removed after using the objective m(a simple wipe > with lens tissue will suffice), gravity will eventually cause the oil to > run down the spring loaded front lens barrel into the objective. Over time > enough oil will build up in the objective, so that it can find its way > between the inner lens elements. > > The other scenario is if the front lens of the objective has been crashed > into a specimen holder or the underside of a x-y stage. This can cause the > seal around the front lens to develop a microcrack and the immersion oil > will find its way into the objective. > > Either of the above damage is usually repairable by the manufacturer, but > not under warranty. > > I have seen people use a piece of a pipe cleaner wrapped around the > objective to catch the excess oil, but you need to remember to change it > regularly. > > David J. Claypool > Digital Imaging Product Manager > Micro Video Instruments > Office: 800-875-2041 x5221 > Cell: 603-809-5342 > [hidden email] > > > On Apr 4, 2016, at 11:16 AM, Zdenek Svindrych <[hidden email]> wrote: > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > > > Hi. > > To me it seems very unlikely someone would do that on purpose. It's more > > likely that some greenhorn user misunderstood the sentence "don't forget > to > > apply oil to the oil immersion objective" or something like that in your > > manual (I'm sure you don't specifically note that it's not necessary to > > unscrew the objective from the turret to do that :-). > > But seriously, if the oil got inside the objective through the front lens > > seal and found its way to the last lens, the lens is probably full of oil > > and can't be fixed. You may try to clean it and test whether it performs > > well, or ask the vendor whether they can check it for you (and ask how > much > > it would cost!). If you have service contract, they might do it for you > for > > free, you might even be covered by some sort of warranty. > > > > I doubt oil could get there from the lens threads through capillary > action, > > and oil is not superfluid... > > > > Good Luck! > > > > zdenek > > > > > > ---------- Původní zpráva ---------- > > Od: Gary Laevsky <[hidden email]> > > Komu: [hidden email] > > Datum: 4. 4. 2016 10:39:50 > > Předmět: Re: Oil on back focal plane of objective > > > > "***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Hi All, > > > > To be clear, if oil was leaking down the barrel of the objective, and > ended > > up on the center of the back focal plan, there would be a trail from the > > edge, right? > > > > This was a solitary drop on the center of the back. > > > > I'm trying to find someone that will correct me from thinking this was a > > malicious act. > > > > Best, > > > > gary > > > > On Mon, Apr 4, 2016 at 5:48 AM, Gary Laevsky <[hidden email]> > > wrote: > > > >> Hi All, > >> > >> Trying not to be cynical here, but, can anybody please explain to me if > >> there is a way for a drop of oil to end up on the center of the back > focal > >> plane of an objective (oil immersion objective) on an inverted scope. No > >> smear from the side. > >> > >> Thanks. > >> > >> -- > >> Best, > >> > >> Gary Laevsky, Ph.D. > >> Director, Confocal Imaging Facility > >> Nikon Center of Excellence > >> Dept. of Molecular Biology > >> Washington Rd. > >> Princeton University > >> Princeton, New Jersey, 08544-1014 > >> (O) 609 258 5432 > >> (C) 508 507 1310 > > > > > > > > -- > > Best, > > > > Gary Laevsky, Ph.D. > > Director, Confocal Imaging Facility > > Nikon Center of Excellence > > Dept. of Molecular Biology > > Washington Rd. > > Princeton University > > Princeton, New Jersey, 08544-1014 > > (O) 609 258 5432 > > (C) 508 507 1310" > |
![]() ![]() |
Christian Elowsky |
![]() |
Craig,
You just made my day, if not month with this suggestion! Thank you! Christian -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau Sent: Monday, April 04, 2016 11:54 AM To: [hidden email] Subject: Re: Oil on back focal plane of objective ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I still stand by my recommendation of cloth hair scrunchies on the objectives. It catches any oil drips running down the side of the objective on inverted scopes. It won't save you from oil leaking into the spring-loaded tips but it at least keeps it from soaking into the optics under the objective. You can even use them for color coding! http://previews.123rf.com/images/belchonock/belchonock1108/belchonock110800172/10220322-hair-scrunchies-on-brown-background-Stock-Photo.jpg Craig Brideau On Mon, Apr 4, 2016 at 9:31 AM, David Claypool <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > To All, > > This is a somewhat common problem with oil immersion objectives on > inverted microscopes and is not limited to any one brand or manufacturer. > > If excess oil is not removed after using the objective m(a simple wipe > with lens tissue will suffice), gravity will eventually cause the oil > to run down the spring loaded front lens barrel into the objective. > Over time enough oil will build up in the objective, so that it can > find its way between the inner lens elements. > > The other scenario is if the front lens of the objective has been > crashed into a specimen holder or the underside of a x-y stage. This > can cause the seal around the front lens to develop a microcrack and > the immersion oil will find its way into the objective. > > Either of the above damage is usually repairable by the manufacturer, > but not under warranty. > > I have seen people use a piece of a pipe cleaner wrapped around the > objective to catch the excess oil, but you need to remember to change > it regularly. > > David J. Claypool > Digital Imaging Product Manager > Micro Video Instruments > Office: 800-875-2041 x5221 > Cell: 603-809-5342 > [hidden email] > > > On Apr 4, 2016, at 11:16 AM, Zdenek Svindrych <[hidden email]> wrote: > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > > > Hi. > > To me it seems very unlikely someone would do that on purpose. It's > > more likely that some greenhorn user misunderstood the sentence > > "don't forget > to > > apply oil to the oil immersion objective" or something like that in > > your manual (I'm sure you don't specifically note that it's not > > necessary to unscrew the objective from the turret to do that :-). > > But seriously, if the oil got inside the objective through the front > > lens seal and found its way to the last lens, the lens is probably > > full of oil and can't be fixed. You may try to clean it and test > > whether it performs well, or ask the vendor whether they can check > > it for you (and ask how > much > > it would cost!). If you have service contract, they might do it for > > you > for > > free, you might even be covered by some sort of warranty. > > > > I doubt oil could get there from the lens threads through capillary > action, > > and oil is not superfluid... > > > > Good Luck! > > > > zdenek > > > > > > ---------- Původní zpráva ---------- > > Od: Gary Laevsky <[hidden email]> > > Komu: [hidden email] > > Datum: 4. 4. 2016 10:39:50 > > Předmět: Re: Oil on back focal plane of objective > > > > "***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Hi All, > > > > To be clear, if oil was leaking down the barrel of the objective, > > and > ended > > up on the center of the back focal plan, there would be a trail from > > the edge, right? > > > > This was a solitary drop on the center of the back. > > > > I'm trying to find someone that will correct me from thinking this > > was a malicious act. > > > > Best, > > > > gary > > > > On Mon, Apr 4, 2016 at 5:48 AM, Gary Laevsky > > <[hidden email]> > > wrote: > > > >> Hi All, > >> > >> Trying not to be cynical here, but, can anybody please explain to > >> me if there is a way for a drop of oil to end up on the center of > >> the back > focal > >> plane of an objective (oil immersion objective) on an inverted > >> scope. No smear from the side. > >> > >> Thanks. > >> > >> -- > >> Best, > >> > >> Gary Laevsky, Ph.D. > >> Director, Confocal Imaging Facility Nikon Center of Excellence > >> Dept. of Molecular Biology Washington Rd. > >> Princeton University > >> Princeton, New Jersey, 08544-1014 > >> (O) 609 258 5432 > >> (C) 508 507 1310 > > > > > > > > -- > > Best, > > > > Gary Laevsky, Ph.D. > > Director, Confocal Imaging Facility > > Nikon Center of Excellence > > Dept. of Molecular Biology > > Washington Rd. > > Princeton University > > Princeton, New Jersey, 08544-1014 > > (O) 609 258 5432 > > (C) 508 507 1310" > |
![]() ![]() |
Craig Brideau |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Glad you liked it! Just remember to change them periodically, they can get quite messy over time! Craig On Mon, Apr 4, 2016 at 11:11 AM, Christian Elowsky <[hidden email]> wrote: > Craig, > > You just made my day, if not month with this suggestion! > > Thank you! > > Christian > > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Craig Brideau > Sent: Monday, April 04, 2016 11:54 AM > To: [hidden email] > Subject: Re: Oil on back focal plane of objective > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I still stand by my recommendation of cloth hair scrunchies on the > objectives. It catches any oil drips running down the side of the objective > on inverted scopes. It won't save you from oil leaking into the > spring-loaded tips but it at least keeps it from soaking into the optics > under the objective. You can even use them for color coding! > > http://previews.123rf.com/images/belchonock/belchonock1108/belchonock110800172/10220322-hair-scrunchies-on-brown-background-Stock-Photo.jpg > > Craig Brideau > > On Mon, Apr 4, 2016 at 9:31 AM, David Claypool <[hidden email]> wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > To All, > > > > This is a somewhat common problem with oil immersion objectives on > > inverted microscopes and is not limited to any one brand or manufacturer. > > > > If excess oil is not removed after using the objective m(a simple wipe > > with lens tissue will suffice), gravity will eventually cause the oil > > to run down the spring loaded front lens barrel into the objective. > > Over time enough oil will build up in the objective, so that it can > > find its way between the inner lens elements. > > > > The other scenario is if the front lens of the objective has been > > crashed into a specimen holder or the underside of a x-y stage. This > > can cause the seal around the front lens to develop a microcrack and > > the immersion oil will find its way into the objective. > > > > Either of the above damage is usually repairable by the manufacturer, > > but not under warranty. > > > > I have seen people use a piece of a pipe cleaner wrapped around the > > objective to catch the excess oil, but you need to remember to change > > it regularly. > > > > David J. Claypool > > Digital Imaging Product Manager > > Micro Video Instruments > > Office: 800-875-2041 x5221 > > Cell: 603-809-5342 > > [hidden email] > > > > > On Apr 4, 2016, at 11:16 AM, Zdenek Svindrych <[hidden email]> > wrote: > > > > > > ***** > > > To join, leave or search the confocal microscopy listserv, go to: > > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > > Post images on http://www.imgur.com and include the link in your > > posting. > > > ***** > > > > > > > > > Hi. > > > To me it seems very unlikely someone would do that on purpose. It's > > > more likely that some greenhorn user misunderstood the sentence > > > "don't forget > > to > > > apply oil to the oil immersion objective" or something like that in > > > your manual (I'm sure you don't specifically note that it's not > > > necessary to unscrew the objective from the turret to do that :-). > > > But seriously, if the oil got inside the objective through the front > > > lens seal and found its way to the last lens, the lens is probably > > > full of oil and can't be fixed. You may try to clean it and test > > > whether it performs well, or ask the vendor whether they can check > > > it for you (and ask how > > much > > > it would cost!). If you have service contract, they might do it for > > > you > > for > > > free, you might even be covered by some sort of warranty. > > > > > > I doubt oil could get there from the lens threads through capillary > > action, > > > and oil is not superfluid... > > > > > > Good Luck! > > > > > > zdenek > > > > > > > > > ---------- Původní zpráva ---------- > > > Od: Gary Laevsky <[hidden email]> > > > Komu: [hidden email] > > > Datum: 4. 4. 2016 10:39:50 > > > Předmět: Re: Oil on back focal plane of objective > > > > > > "***** > > > To join, leave or search the confocal microscopy listserv, go to: > > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > > Post images on http://www.imgur.com and include the link in your > > posting. > > > ***** > > > > > > Hi All, > > > > > > To be clear, if oil was leaking down the barrel of the objective, > > > and > > ended > > > up on the center of the back focal plan, there would be a trail from > > > the edge, right? > > > > > > This was a solitary drop on the center of the back. > > > > > > I'm trying to find someone that will correct me from thinking this > > > was a malicious act. > > > > > > Best, > > > > > > gary > > > > > > On Mon, Apr 4, 2016 at 5:48 AM, Gary Laevsky > > > <[hidden email]> > > > wrote: > > > > > >> Hi All, > > >> > > >> Trying not to be cynical here, but, can anybody please explain to > > >> me if there is a way for a drop of oil to end up on the center of > > >> the back > > focal > > >> plane of an objective (oil immersion objective) on an inverted > > >> scope. No smear from the side. > > >> > > >> Thanks. > > >> > > >> -- > > >> Best, > > >> > > >> Gary Laevsky, Ph.D. > > >> Director, Confocal Imaging Facility Nikon Center of Excellence > > >> Dept. of Molecular Biology Washington Rd. > > >> Princeton University > > >> Princeton, New Jersey, 08544-1014 > > >> (O) 609 258 5432 > > >> (C) 508 507 1310 > > > > > > > > > > > > -- > > > Best, > > > > > > Gary Laevsky, Ph.D. > > > Director, Confocal Imaging Facility > > > Nikon Center of Excellence > > > Dept. of Molecular Biology > > > Washington Rd. > > > Princeton University > > > Princeton, New Jersey, 08544-1014 > > > (O) 609 258 5432 > > > (C) 508 507 1310" > > > |
![]() ![]() |
Eric Griffis |
![]() |
In reply to this post by Craig Brideau
I second this suggestion. They are cheap, easy to replace, and effective. Plus, when you change them, you can display the sodden old ones to show users that they are putting too much oil on the objectives. If you're still getting oil infiltration with these in place, you know you need some remedial user training.
Eric Centre for Gene Regulation and Expression College of Life Sciences University of Dundee MSI/WTB/JBC Complex Dow Street Dundee DD1 5EH United KIngdom +44 (0)1382 385118 [hidden email]<mailto:[hidden email]> http://www.lifesci.dundee.ac.uk/groups/eric_griffis/GriffisLab/index.html On 4 Apr 2016, at 17:54, Craig Brideau <[hidden email]<mailto:[hidden email]>> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I still stand by my recommendation of cloth hair scrunchies on the objectives. It catches any oil drips running down the side of the objective on inverted scopes. It won't save you from oil leaking into the spring-loaded tips but it at least keeps it from soaking into the optics under the objective. You can even use them for color coding! http://previews.123rf.com/images/belchonock/belchonock1108/belchonock110800172/10220322-hair-scrunchies-on-brown-background-Stock-Photo.jpg Craig Brideau On Mon, Apr 4, 2016 at 9:31 AM, David Claypool <[hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** To All, This is a somewhat common problem with oil immersion objectives on inverted microscopes and is not limited to any one brand or manufacturer. If excess oil is not removed after using the objective m(a simple wipe with lens tissue will suffice), gravity will eventually cause the oil to run down the spring loaded front lens barrel into the objective. Over time enough oil will build up in the objective, so that it can find its way between the inner lens elements. The other scenario is if the front lens of the objective has been crashed into a specimen holder or the underside of a x-y stage. This can cause the seal around the front lens to develop a microcrack and the immersion oil will find its way into the objective. Either of the above damage is usually repairable by the manufacturer, but not under warranty. I have seen people use a piece of a pipe cleaner wrapped around the objective to catch the excess oil, but you need to remember to change it regularly. David J. Claypool Digital Imaging Product Manager Micro Video Instruments Office: 800-875-2041 x5221 Cell: 603-809-5342 [hidden email] On Apr 4, 2016, at 11:16 AM, Zdenek Svindrych <[hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi. To me it seems very unlikely someone would do that on purpose. It's more likely that some greenhorn user misunderstood the sentence "don't forget to apply oil to the oil immersion objective" or something like that in your manual (I'm sure you don't specifically note that it's not necessary to unscrew the objective from the turret to do that :-). But seriously, if the oil got inside the objective through the front lens seal and found its way to the last lens, the lens is probably full of oil and can't be fixed. You may try to clean it and test whether it performs well, or ask the vendor whether they can check it for you (and ask how much it would cost!). If you have service contract, they might do it for you for free, you might even be covered by some sort of warranty. I doubt oil could get there from the lens threads through capillary action, and oil is not superfluid... Good Luck! zdenek ---------- Původní zpráva ---------- Od: Gary Laevsky <[hidden email]> Komu: [hidden email] Datum: 4. 4. 2016 10:39:50 Předmět: Re: Oil on back focal plane of objective "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi All, To be clear, if oil was leaking down the barrel of the objective, and ended up on the center of the back focal plan, there would be a trail from the edge, right? This was a solitary drop on the center of the back. I'm trying to find someone that will correct me from thinking this was a malicious act. Best, gary On Mon, Apr 4, 2016 at 5:48 AM, Gary Laevsky <[hidden email]> wrote: Hi All, Trying not to be cynical here, but, can anybody please explain to me if there is a way for a drop of oil to end up on the center of the back focal plane of an objective (oil immersion objective) on an inverted scope. No smear from the side. Thanks. -- Best, Gary Laevsky, Ph.D. Director, Confocal Imaging Facility Nikon Center of Excellence Dept. of Molecular Biology Washington Rd. Princeton University Princeton, New Jersey, 08544-1014 (O) 609 258 5432 (C) 508 507 1310 -- Best, Gary Laevsky, Ph.D. Director, Confocal Imaging Facility Nikon Center of Excellence Dept. of Molecular Biology Washington Rd. Princeton University Princeton, New Jersey, 08544-1014 (O) 609 258 5432 (C) 508 507 1310" The University of Dundee is a registered Scottish Charity, No: SC015096 |
![]() ![]() |
Quintanar, Loren |
![]() |
In reply to this post by Christian Elowsky
I doubt this was an intentional act by someone. I understand from Zeiss reps and other sources that oil leakage into the objective barrel usually happens over time. However, I can imagine that a LOT of oil getting into the side crevices would show up in the bottom lens fairly quickly, but I cannot imagine someone would go out of their way to pour oil into the side crevices (Unless they are a knowledgeable microscopy bandit that knows how to damage lenses on purpose).
We use little hair "scrunchies" to catch excess oil on our inverted microscopes as well (I even rotate colors to celebrate relevant Holidays). Besides regularly cleaning the front lenses after use, I usually take off the scrunchies about once a week and thoroughly soak/wipe them with 95% ethanol. Maybe once a month I replace them, as they do get nasty. I am sometimes surprised by the translucence of the oil that makes its way on the side of the objective. Even if I see no oil "trail," I can usually feel the oil. I rarely let oil sit on a lens for more than a few hours (i.e. avoid overnight oil sitting). As always, the best way to prevent any damage is to regularly remind users that they need way less oil than they think! It seems possible that there could be invisible damage in/around the front lens, which caused the oil to leak in. Then again, I have a hard time wrapping my head around how it got to the bottom lens and possibly surpassing other optics inside. Loren -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christian Elowsky Sent: Monday, April 04, 2016 10:12 AM To: [hidden email] Subject: Re: Oil on back focal plane of objective Craig, You just made my day, if not month with this suggestion! Thank you! Christian -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau Sent: Monday, April 04, 2016 11:54 AM To: [hidden email] Subject: Re: Oil on back focal plane of objective ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I still stand by my recommendation of cloth hair scrunchies on the objectives. It catches any oil drips running down the side of the objective on inverted scopes. It won't save you from oil leaking into the spring-loaded tips but it at least keeps it from soaking into the optics under the objective. You can even use them for color coding! http://previews.123rf.com/images/belchonock/belchonock1108/belchonock110800172/10220322-hair-scrunchies-on-brown-background-Stock-Photo.jpg Craig Brideau On Mon, Apr 4, 2016 at 9:31 AM, David Claypool <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > To All, > > This is a somewhat common problem with oil immersion objectives on > inverted microscopes and is not limited to any one brand or manufacturer. > > If excess oil is not removed after using the objective m(a simple wipe > with lens tissue will suffice), gravity will eventually cause the oil > to run down the spring loaded front lens barrel into the objective. > Over time enough oil will build up in the objective, so that it can > find its way between the inner lens elements. > > The other scenario is if the front lens of the objective has been > crashed into a specimen holder or the underside of a x-y stage. This > can cause the seal around the front lens to develop a microcrack and > the immersion oil will find its way into the objective. > > Either of the above damage is usually repairable by the manufacturer, > but not under warranty. > > I have seen people use a piece of a pipe cleaner wrapped around the > objective to catch the excess oil, but you need to remember to change > it regularly. > > David J. Claypool > Digital Imaging Product Manager > Micro Video Instruments > Office: 800-875-2041 x5221 > Cell: 603-809-5342 > [hidden email] > > > On Apr 4, 2016, at 11:16 AM, Zdenek Svindrych <[hidden email]> wrote: > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > > > Hi. > > To me it seems very unlikely someone would do that on purpose. It's > > more likely that some greenhorn user misunderstood the sentence > > "don't forget > to > > apply oil to the oil immersion objective" or something like that in > > your manual (I'm sure you don't specifically note that it's not > > necessary to unscrew the objective from the turret to do that :-). > > But seriously, if the oil got inside the objective through the front > > lens seal and found its way to the last lens, the lens is probably > > full of oil and can't be fixed. You may try to clean it and test > > whether it performs well, or ask the vendor whether they can check > > it for you (and ask how > much > > it would cost!). If you have service contract, they might do it for > > you > for > > free, you might even be covered by some sort of warranty. > > > > I doubt oil could get there from the lens threads through capillary > action, > > and oil is not superfluid... > > > > Good Luck! > > > > zdenek > > > > > > ---------- Původní zpráva ---------- > > Od: Gary Laevsky <[hidden email]> > > Komu: [hidden email] > > Datum: 4. 4. 2016 10:39:50 > > Předmět: Re: Oil on back focal plane of objective > > > > "***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Hi All, > > > > To be clear, if oil was leaking down the barrel of the objective, > > and > ended > > up on the center of the back focal plan, there would be a trail from > > the edge, right? > > > > This was a solitary drop on the center of the back. > > > > I'm trying to find someone that will correct me from thinking this > > was a malicious act. > > > > Best, > > > > gary > > > > On Mon, Apr 4, 2016 at 5:48 AM, Gary Laevsky > > <[hidden email]> > > wrote: > > > >> Hi All, > >> > >> Trying not to be cynical here, but, can anybody please explain to > >> me if there is a way for a drop of oil to end up on the center of > >> the back > focal > >> plane of an objective (oil immersion objective) on an inverted > >> scope. No smear from the side. > >> > >> Thanks. > >> > >> -- > >> Best, > >> > >> Gary Laevsky, Ph.D. > >> Director, Confocal Imaging Facility Nikon Center of Excellence > >> Dept. of Molecular Biology Washington Rd. > >> Princeton University > >> Princeton, New Jersey, 08544-1014 > >> (O) 609 258 5432 > >> (C) 508 507 1310 > > > > > > > > -- > > Best, > > > > Gary Laevsky, Ph.D. > > Director, Confocal Imaging Facility > > Nikon Center of Excellence > > Dept. of Molecular Biology > > Washington Rd. > > Princeton University > > Princeton, New Jersey, 08544-1014 > > (O) 609 258 5432 > > (C) 508 507 1310" > --------------------------------------------------------------------- *SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. (fpc5p) --------------------------------------------------------------------- |
![]() ![]() |
Ralf Palmisano |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi, I am not cynical and I don't mean to be mean to anyone here. I will just name facts, eventually ask a "rhetorical" question and leave the rest up to you. - A nice solitary drop on the backfocal plane of an inverted microscope can physically only happen, if the turret is not being used for quite a while. Was this the case? - If no smear is on the outside, it must have been making its way through the objectives inner parts and lenses. - If the oil went down the outside, no matter what better how long you wait, there will always be a smear trail. Only just not if you have not used it for about a year or two. Did you? - How does one realise there is a drop of oil on the backfocal plane? I would guess because a user or member of staff realises you only get a a distorted image, whatever you do? - If such an amount of oil leaks through the inner parts of an objective lens you will already realise a visible impairment of your light passing through the objective lens long before oil getting to the back focal plane, also quite early leading to a distorted image. Unless all your users are blind or have no clue what they are imaging or looking at. You can easily check the front lens of an objective using a stereo or simple zoom microscope. If it leaked through the anti-shock you always easily can feel that the anti-shock is no longer moving smoothly - apart from the fact you managed to get solid dirt in there. Therefore you have to strongly consider some maleficent user in your facility. In about 17 years of running four different imaging facilities, with probably more than a thousand users, I have experienced much worse deliberately taken action by some weird, disappointed or mistreated by their PIs users. That is my unconvenient expertise. -- Ralf Palmisano Head - Optical Imaging Centre Erlangen Hartmannstr. 14 91052 Erlangen, Germany +49-9131-85-64300 (Office) +49-9131-85-64301 (Secretary) +49-9131-85-64302 (Fax) www.oice.uni-erlangen.de |
![]() |
In reply to this post by David Claypool
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** We have not had a problem with oil getting inside an objective so that it compromises resolution. However, it can get on the outside of the inner barrel and run down inside the microscope body where it gets on dichroics, other glass surfaces, or onto microschips where is destroys them We've also had problems with salty water (PBS? Media?) dripping inside the scopes and destroying mechanical parts. ========================================================================= Michael Cammer, Microscopy Core & Skirball Institute, NYU Langone Medical Center Cell: 914-309-3270 ** Office: Skirball 2nd Floor main office, back right ** http://ocs.med.nyu.edu/microscopy & http://microscopynotes.com/ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of David Claypool Sent: Monday, April 04, 2016 11:32 AM To: [hidden email] Subject: Re: Oil on back focal plane of objective ***** To join, leave or search the confocal microscopy listserv, go to: https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi-2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=CwIFBA&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=0_k5e3G-Hn49IB14k9dJGFlZc5DXSvZO_7gzJFinUEk&s=UVMjVrVCxdXrHEXBV1-fiyUfB4zfb6efRryqxjLCy08&e= Post images on https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=CwIFBA&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=0_k5e3G-Hn49IB14k9dJGFlZc5DXSvZO_7gzJFinUEk&s=yfwCDBdHQKzrvPQtmK0Le5aEHd2TxzRDXAi69udTgs8&e= and include the link in your posting. ***** To All, This is a somewhat common problem with oil immersion objectives on inverted microscopes and is not limited to any one brand or manufacturer. If excess oil is not removed after using the objective m(a simple wipe with lens tissue will suffice), gravity will eventually cause the oil to run down the spring loaded front lens barrel into the objective. Over time enough oil will build up in the objective, so that it can find its way between the inner lens elements. The other scenario is if the front lens of the objective has been crashed into a specimen holder or the underside of a x-y stage. This can cause the seal around the front lens to develop a microcrack and the immersion oil will find its way into the objective. Either of the above damage is usually repairable by the manufacturer, but not under warranty. I have seen people use a piece of a pipe cleaner wrapped around the objective to catch the excess oil, but you need to remember to change it regularly. David J. Claypool Digital Imaging Product Manager Micro Video Instruments Office: 800-875-2041 x5221 Cell: 603-809-5342 [hidden email] > On Apr 4, 2016, at 11:16 AM, Zdenek Svindrych <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi- > 2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=CwIFBA&c=j5oPpO0eBH1iio48DtsedbOB > Gmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=0_ > k5e3G-Hn49IB14k9dJGFlZc5DXSvZO_7gzJFinUEk&s=UVMjVrVCxdXrHEXBV1-fiyUfB4 > zfb6efRryqxjLCy08&e= Post images on > https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=CwIFBA&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=0_k5e3G-Hn49IB14k9dJGFlZc5DXSvZO_7gzJFinUEk&s=yfwCDBdHQKzrvPQtmK0Le5aEHd2TxzRDXAi69udTgs8&e= and include the link in your posting. > ***** > > > Hi. > To me it seems very unlikely someone would do that on purpose. It's > more likely that some greenhorn user misunderstood the sentence "don't > forget to apply oil to the oil immersion objective" or something like > that in your manual (I'm sure you don't specifically note that it's > not necessary to unscrew the objective from the turret to do that :-). > But seriously, if the oil got inside the objective through the front > lens seal and found its way to the last lens, the lens is probably > full of oil and can't be fixed. You may try to clean it and test > whether it performs well, or ask the vendor whether they can check it > for you (and ask how much it would cost!). If you have service > contract, they might do it for you for free, you might even be covered by some sort of warranty. > > I doubt oil could get there from the lens threads through capillary > action, and oil is not superfluid... > > Good Luck! > > zdenek > > > ---------- Původní zpráva ---------- > Od: Gary Laevsky <[hidden email]> > Komu: [hidden email] > Datum: 4. 4. 2016 10:39:50 > Předmět: Re: Oil on back focal plane of objective > > "***** > To join, leave or search the confocal microscopy listserv, go to: > https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.umn.edu_cgi- > 2Dbin_wa-3FA0-3Dconfocalmicroscopy&d=CwIFBA&c=j5oPpO0eBH1iio48DtsedbOB > Gmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=0_ > k5e3G-Hn49IB14k9dJGFlZc5DXSvZO_7gzJFinUEk&s=UVMjVrVCxdXrHEXBV1-fiyUfB4 > zfb6efRryqxjLCy08&e= Post images on > https://urldefense.proofpoint.com/v2/url?u=http-3A__www.imgur.com&d=CwIFBA&c=j5oPpO0eBH1iio48DtsedbOBGmuw5jHLjgvtN2r4ehE&r=oU_05LztNstAydlbm5L5GDu_vAdjXk3frDLx_CqKkuo&m=0_k5e3G-Hn49IB14k9dJGFlZc5DXSvZO_7gzJFinUEk&s=yfwCDBdHQKzrvPQtmK0Le5aEHd2TxzRDXAi69udTgs8&e= and include the link in your posting. > ***** > > Hi All, > > To be clear, if oil was leaking down the barrel of the objective, and > ended up on the center of the back focal plan, there would be a trail > from the edge, right? > > This was a solitary drop on the center of the back. > > I'm trying to find someone that will correct me from thinking this was > a malicious act. > > Best, > > gary > > On Mon, Apr 4, 2016 at 5:48 AM, Gary Laevsky > <[hidden email]> > wrote: > >> Hi All, >> >> Trying not to be cynical here, but, can anybody please explain to me >> if there is a way for a drop of oil to end up on the center of the >> back focal plane of an objective (oil immersion objective) on an >> inverted scope. No smear from the side. >> >> Thanks. >> >> -- >> Best, >> >> Gary Laevsky, Ph.D. >> Director, Confocal Imaging Facility >> Nikon Center of Excellence >> Dept. of Molecular Biology >> Washington Rd. >> Princeton University >> Princeton, New Jersey, 08544-1014 >> (O) 609 258 5432 >> (C) 508 507 1310 > > > > -- > Best, > > Gary Laevsky, Ph.D. > Director, Confocal Imaging Facility > Nikon Center of Excellence > Dept. of Molecular Biology > Washington Rd. > Princeton University > Princeton, New Jersey, 08544-1014 > (O) 609 258 5432 > (C) 508 507 1310" ------------------------------------------------------------ This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email. ================================= |
![]() |
In reply to this post by Zdenek Svindrych-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** This is not a solution but it is an example of how a misunderstanding can foul up a scope. In the early 80s I was called to a major hospital dept. of immunology to check out a Zeiss Photo 3 fluorescence scope. Okay, okay, it was before the digital age, I know! The reason the scope did not work and would not form an image was that a new user was told to use the "immersion scope" to read TB slides. The FNG (fabulous new guy) dumped an entire bottle of immersion oil into the camera port!!! It ended up that the new guy did not understand English well enough to understand how to use the scope. That's a true story and a sad example of what can happen to a scope in a lab. Dan On 4/4/2016 11:15 AM, Zdenek Svindrych wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > > Hi. > To me it seems very unlikely someone would do that on purpose. It's more > likely that some greenhorn user misunderstood the sentence "don't forget to > apply oil to the oil immersion objective" or something like that in your > manual (I'm sure you don't specifically note that it's not necessary to > unscrew the objective from the turret to do that :-). > But seriously, if the oil got inside the objective through the front lens > seal and found its way to the last lens, the lens is probably full of oil > and can't be fixed. You may try to clean it and test whether it performs > well, or ask the vendor whether they can check it for you (and ask how much > it would cost!). If you have service contract, they might do it for you for > free, you might even be covered by some sort of warranty. > > I doubt oil could get there from the lens threads through capillary action, > and oil is not superfluid... > > Good Luck! > > zdenek > > > ---------- Původní zpráva ---------- > Od: Gary Laevsky <[hidden email]> > Komu: [hidden email] > Datum: 4. 4. 2016 10:39:50 > Předmět: Re: Oil on back focal plane of objective > > "***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi All, > > To be clear, if oil was leaking down the barrel of the objective, and ended > up on the center of the back focal plan, there would be a trail from the > edge, right? > > This was a solitary drop on the center of the back. > > I'm trying to find someone that will correct me from thinking this was a > malicious act. > > Best, > > gary > > On Mon, Apr 4, 2016 at 5:48 AM, Gary Laevsky <[hidden email]> > wrote: > >> Hi All, >> >> Trying not to be cynical here, but, can anybody please explain to me if >> there is a way for a drop of oil to end up on the center of the back focal >> plane of an objective (oil immersion objective) on an inverted scope. No >> smear from the side. >> >> Thanks. >> >> -- >> Best, >> >> Gary Laevsky, Ph.D. >> Director, Confocal Imaging Facility >> Nikon Center of Excellence >> Dept. of Molecular Biology >> Washington Rd. >> Princeton University >> Princeton, New Jersey, 08544-1014 >> (O) 609 258 5432 >> (C) 508 507 1310 >> > > -- Dan Focht Bioptechs Inc. V: (724)282-7145 www.bioptechs.com |
Free forum by Nabble | Edit this page |