![]() ![]() |
Linda Barthel |
![]() |
Seems to be a simple enough question-but having trouble getting a simple
answer. Operating a Leica SP5 Confocal microscope, software: LAS-AF ver. 2.0.0 software. How do we determine the optical sections thickness of our images? This must be a simple oversight on our part. Thanks, Linda Barthel Gregg Sobocinski University of Michigan Dept of MCDB Ann Arbor |
![]() ![]() |
James Pawley |
![]() |
>Seems to be a simple enough question-but having trouble getting a simple
>answer. Operating a Leica SP5 Confocal microscope, software: LAS-AF ver. >2.0.0 software. How do we determine the optical sections thickness of our >images? This must be a simple oversight on our part. >Thanks, >Linda Barthel >Gregg Sobocinski >University of Michigan >Dept of MCDB >Ann Arbor Hi Linda, Optical thickness is the same as z-resolution. It varies directly with wavelength and inversely with (NA) squared. So the actual number depends... Besides this Abbe equation-like dependence just noted, z-resolution can also be strongly degraded by the presence of spherical aberration, the extent which the full numerical aperture is filled with laser light, and how big your pinhole diameter is. In simple terms, SA greatly extends the point-spread function in the z-direction, even when it presence is hardly evident when looking at an XY image. Short version, unless you have taken precautions to detect and correct SA, the actual z-resolution of any image of a living cell is likely to be much worse (at least 2x) than that which you would calculate from the equations. Cheers, Jim Pawley -- ********************************************************************************* Prof. James B. Pawley, Ph. 608-263-3147 Room 223, Zoology Research Building, FAX 608-265-5315 1117 Johnson Ave., Madison, WI, 53706 [hidden email] 3D Microscopy of Living Cells Course, June 12-24, 2010, UBC, Vancouver Canada Info: http://www.3dcourse.ubc.ca/ Applications still being accepted "If it ain't diffraction, it must be statistics." Anon. |
![]() ![]() |
Patrick Van Oostveldt |
![]() |
In reply to this post by Linda Barthel
If you really want to determine the thickness of a section you should
take account of the fact that the refractive index of the medium can change the estimated thickness. Cfr. Paper of Visser, Oud and Brakenhoff in the early days of CSLM. Just google to "visser refractive index" and you will find the pdf print. My experience is that fixation can also affect effective thickness of the section due to shrinking or swelling. A test is possible by using a fluorescent dye mounted under a coverglass with specific spacers and measure the number of steps necessary to image the whole object. A simple test is the use of some fluorescentive index and axial ...Quoting "Linda Barthel" <[hidden email]>: textile fiber of PE which is round and where you can directly check if XY diameter corresponds to XZ distances. Bye Patrick Van Oostveldt > Seems to be a simple enough question-but having trouble getting a simple > answer. Operating a Leica SP5 Confocal microscope, software: LAS-AF ver. > 2.0.0 software. How do we determine the optical sections thickness of our > images? This must be a simple oversight on our part. > Thanks, > Linda Barthel > Gregg Sobocinski > University of Michigan > Dept of MCDB > Ann Arbor > -- Dep. Moleculaire Biotechnologie Coupure links 653 B 9000 GENT tel 09 264 5969 fax 09 264 6219 |
![]() ![]() |
Ignatius, Mike-2 |
![]() |
For specific "spacers" our 15 micron fluorescent beads were designed for CLSM based Z (depth) calibration. We also have 6 micron ones if you prefer.
These also show the effect of color and RI mismatch on Z depth accuracy. Simple protocol included. FocalCheck™ microspheres, 15 µm, fluorescent green ring stain⁄dark red throughout Cat. No. F-7238 Or FocalCheck™ microspheres, 15 µm, fluorescent dark-red ring stain⁄green throughout Cat. No. F-7239 Kind Regards, Mike Ignatius Molecular Probes/Life Technologies -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Patrick Van Oostveldt Sent: Tuesday, May 11, 2010 10:24 AM To: [hidden email] Subject: Re: Optical Section Thickness-Leica SP5 If you really want to determine the thickness of a section you should take account of the fact that the refractive index of the medium can change the estimated thickness. Cfr. Paper of Visser, Oud and Brakenhoff in the early days of CSLM. Just google to "visser refractive index" and you will find the pdf print. My experience is that fixation can also affect effective thickness of the section due to shrinking or swelling. A test is possible by using a fluorescent dye mounted under a coverglass with specific spacers and measure the number of steps necessary to image the whole object. A simple test is the use of some fluorescentive index and axial ...Quoting "Linda Barthel" <[hidden email]>: textile fiber of PE which is round and where you can directly check if XY diameter corresponds to XZ distances. Bye Patrick Van Oostveldt > Seems to be a simple enough question-but having trouble getting a simple > answer. Operating a Leica SP5 Confocal microscope, software: LAS-AF ver. > 2.0.0 software. How do we determine the optical sections thickness of our > images? This must be a simple oversight on our part. > Thanks, > Linda Barthel > Gregg Sobocinski > University of Michigan > Dept of MCDB > Ann Arbor > -- Dep. Moleculaire Biotechnologie Coupure links 653 B 9000 GENT tel 09 264 5969 fax 09 264 6219 |
![]() |
My inclination is to use subresolution size beads
(< 200 nm) of different colors. As Jim P. pointed out the z-res. is easily twice as bad as the theoretical values. Further, objectives that are nominally of the same high NA (1.4 and above) can yield quite different z-resolutions even factors of 2. In addition, the z-axis PSF can be very asymmetric, which becomes obvious when using sub-res. beads and may be less obvious using very large beads. Making sure you use the correct coverslip thickness and be aware that the z-res. is strongly dependent on the mediums RI. Depending on the amount of SA of you particular objective it will also get worse as you focus deeper into the sample, particularly as you get more than 10 um from the coverslip. Some deconvolution software can at least partially correct for the z-axis spread caused by SA if good calibration data is available. It just depends on how fussy you want to be. Mario >For specific "spacers" our 15 micron fluorescent >beads were designed for CLSM based Z (depth) >calibration. We also have 6 micron ones if you >prefer. > >These also show the effect of color and RI >mismatch on Z depth accuracy. Simple protocol >included. > >FocalCheck microspheres, 15 µm, fluorescent >green ring staindark red throughout >Cat. No. F-7238 >Or >FocalCheck microspheres, 15 µm, fluorescent >dark-red ring staingreen throughout >Cat. No. F-7239 > >Kind Regards, > >Mike Ignatius >Molecular Probes/Life Technologies > >-----Original Message----- >From: Confocal Microscopy List >[mailto:[hidden email]] On >Behalf Of Patrick Van Oostveldt >Sent: Tuesday, May 11, 2010 10:24 AM >To: [hidden email] >Subject: Re: Optical Section Thickness-Leica SP5 > >If you really want to determine the thickness of a section you should >take account of the fact that the refractive index of the medium can >change the estimated thickness. Cfr. Paper of Visser, Oud and >Brakenhoff in the early days of CSLM. Just google to "visser >refractive index" and you will find the pdf print. > >My experience is that fixation can also affect effective thickness of >the section due to shrinking or swelling. > >A test is possible by using a fluorescent dye mounted under a >coverglass with specific spacers and measure the number of steps >necessary to image the whole object. > >A simple test is the use of some fluorescentive index and axial >...Quoting "Linda Barthel" <[hidden email]>: textile fiber of PE >which is round and where you can directly check if XY diameter >corresponds to XZ distances. > >Bye > >Patrick Van Oostveldt > > > > >> Seems to be a simple enough question-but having trouble getting a simple >> answer. Operating a Leica SP5 Confocal microscope, software: LAS-AF ver. >> 2.0.0 software. How do we determine the optical sections thickness of our >> images? This must be a simple oversight on our part. >> Thanks, >> Linda Barthel >> Gregg Sobocinski >> University of Michigan >> Dept of MCDB >> Ann Arbor >> > > > >-- >Dep. Moleculaire Biotechnologie >Coupure links 653 >B 9000 GENT > >tel 09 264 5969 >fax 09 264 6219 -- ________________________________________________________________________________ Mario M. Moronne, Ph.D. [hidden email] [hidden email] |
![]() ![]() |
Julio Vazquez |
![]() |
In reply to this post by Linda Barthel
Hi Linda,
I'm guessing you want to know how to set up optimal z spacing based on your pinhole aperture and objective. I am not familiar with SP5, but I believe when you go to the z stack settings panel, there should be a button named "system optimized" that will match your pinhole aperture and section distance. This will be typically about half of the z resolution (or z focal depth). If you can't figure that out, you may search Google for leica las-af manual or leica sp5 manual, and you will get to some tutorials that may have the desired info. Also, a call or email to your Leica rep should give you the answer. Just ask them where in te hsoftware do you set up the optimal z section spacing or thickness.
-- Julio Vazquez Fred Hutchinson Cancer Research Center 1100 Fairview Ave. N., mailstop DE-512 Seattle, WA 98109-1024 On May 11, 2010, at 9:05 AM, Linda Barthel wrote:
|
![]() |
In reply to this post by Linda Barthel
Hi james
You can ask leica for software version2.2.1, Goodluck Xuelin 发自我的 iPod 在 2010-5-12,0:05,Linda Barthel <[hidden email]> 写到: > Seems to be a simple enough question-but having trouble getting a > simple > answer. Operating a Leica SP5 Confocal microscope, software: LAS-AF > ver. > 2.0.0 software. How do we determine the optical sections thickness > of our > images? This must be a simple oversight on our part. > Thanks, > Linda Barthel > Gregg Sobocinski > University of Michigan > Dept of MCDB > Ann Arbor |
![]() ![]() |
Adrian Smith-6 |
![]() |
In reply to this post by Julio Vazquez
The version of LAS AF that we have (2.2.x) has "optimised" z-slicing in two locations.
One is as Julio describes below, ie in the z-stack dialog. To get the z-step it takes the value reported in the objective information dialog and applies an 2-step Nyquist approximation (so the z-step size is half the z-resolution reported in the objective information pane). The objective dialog seem to use a very "optimistic" formula for z-resolution. The other location is in the extra controls that come up when you click the pinhole radio box (ie disable automatic pinhole settings) - this reveals a new button which brings up a new window with a more conservsative formula with the option to enter the wavelength you are interested in. The calculated thickness changes with the pinhole setting which is really helpful when you have opened up the pinhole for some reason. There is another button to copy this to the z-stack dialog. z-steps set this way are about 2-fold larger than those using the first method. I don't know which is the "correct" forumula but based on what I understand/have heard about z-resolution the older, smaller numbers in the objective dialog seem unrealistic. I can't remember which version (it was sometime after 2.0) added the extra controls in the pinhole dialog. Hope that helps, Regards, Adrian Smith Centenary Institute, Sydney, Australia On 12/05/2010, at 5:05 AM, Julio Vazquez wrote:
|
![]() |
Hi All,
Our University has not historically considered leasing as an option, but we have a number of immediate major equipment needs that might be well-served by this mechanism. Does anyone have experience either 3rd party or directly with the manufacturer in a University environment? Any special issues? Were leasing conditions favorable? Any 3rd party companies that were good to deal with? Any and all comments and suggestions appreciated. I will compile results and share with the list for others out there wondering how to keep-up with the pace of technology and instrumentation. Best Regards, Kirk Kirk J. Czymmek, Ph.D. Associate Professor Department of Biological Sciences Delaware Biotechnology Institute Bio-Imaging Center, Director University of Delaware Newark, DE 19716 |
![]() |
Hi Kirk,
We have had our clients use Heartland Business Credit in the past and they seem to be a pretty good company to deal with. As a general rule of thumb you would get a better interest rate from a third party leasing company vs. a vendor as in general a vendor's cost of capital is higher and therefore there interest rate charge to you would be higher. Hope this helps. Dave -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of kirk Sent: Wednesday, May 12, 2010 3:33 AM To: [hidden email] Subject: Major Equipment Leasing Hi All, Our University has not historically considered leasing as an option, but we have a number of immediate major equipment needs that might be well-served by this mechanism. Does anyone have experience either 3rd party or directly with the manufacturer in a University environment? Any special issues? Were leasing conditions favorable? Any 3rd party companies that were good to deal with? Any and all comments and suggestions appreciated. I will compile results and share with the list for others out there wondering how to keep-up with the pace of technology and instrumentation. Best Regards, Kirk Kirk J. Czymmek, Ph.D. Associate Professor Department of Biological Sciences Delaware Biotechnology Institute Bio-Imaging Center, Director University of Delaware Newark, DE 19716 ==================================================================================== The information contained in this e-mail message is intended only for the personal and confidential use of the recipient(s) named above. If the reader of this message is not the intended recipient or an agent responsible for delivering it to the intended recipient, you are hereby notified that you have received this document in error and that any review, dissemination, distribution, or copying of this message is strictly prohibited. If you have received this communication in error, please notify us immediately by e-mail, and delete the original message. ==================================================================================== |
![]() ![]() |
Gregg Sobocinski |
![]() |
In reply to this post by Adrian Smith-6
Thanks to those who responded to our inquiry about determining
optical thickness of images captured on the Leica SP5. We’re happy with
the responses from both this Listserver community and the Leica support
specialists. Between the Listserver responses, and the helpful response of
Leica, we now have the latest software (which calculates optical thickness for
us) as well as a complete understanding of the variables involved in reaching
that number. With gratitude, ~Gregg Gregg Sobocinski Microscope Imaging Specialist University of Michigan, MCDB Dept. Ann Arbor, Michigan USA From: Confocal Microscopy
List [mailto:[hidden email]] On Behalf Of Adrian Smith The version of LAS AF that we have (2.2.x) has "optimised"
z-slicing in two locations. One is as Julio describes below, ie in the z-stack dialog.
To get the z-step it takes the value reported in the objective information
dialog and applies an 2-step Nyquist approximation (so the z-step size is half
the z-resolution reported in the objective information pane). The
objective dialog seem to use a very "optimistic" formula for
z-resolution. The other location is in the extra controls that come up
when you click the pinhole radio box (ie disable automatic pinhole settings) -
this reveals a new button which brings up a new window with a more
conservsative formula with the option to enter the wavelength you are
interested in. The calculated thickness changes with the pinhole setting which
is really helpful when you have opened up the pinhole for some reason. There is
another button to copy this to the z-stack dialog. z-steps set this way are
about 2-fold larger than those using the first method. I don't know which is the "correct" forumula but
based on what I understand/have heard about z-resolution the older, smaller
numbers in the objective dialog seem unrealistic. I can't remember which version (it was sometime after 2.0)
added the extra controls in the pinhole dialog. Hope that helps, Regards, Adrian Smith Centenary Institute, Sydney, Australia On 12/05/2010, at 5:05 AM, Julio Vazquez wrote:
Hi Linda, I'm guessing you want to know how to set up optimal z
spacing based on your pinhole aperture and objective. I am not familiar with SP5,
but I believe when you go to the z stack settings panel, there should be a
button named "system optimized" that will match your pinhole aperture
and section distance. This will be typically about half of the z resolution (or
z focal depth). If you can't figure that out, you may search Google for leica
las-af manual or leica sp5 manual, and you will get to some tutorials that may
have the desired info. Also, a call or email to your Leica rep should give you
the answer. Just ask them where in te hsoftware do you set up the optimal z
section spacing or thickness. -- Julio Vazquez Fred Hutchinson Cancer
Research Center 1100 Fairview Ave.
N., mailstop DE-512 Seattle, WA 98109-1024 On May 11, 2010, at 9:05 AM, Linda Barthel wrote:
Seems
to be a simple enough question-but having trouble getting a simple answer. Operating a Leica SP5 Confocal
microscope, software: LAS-AF ver. 2.0.0
software. How do we determine
the optical sections thickness of our images? This must be a simple oversight on
our part. Thanks, Linda
Barthel Gregg
Sobocinski University
of Michigan Dept
of MCDB Ann
Arbor |
Free forum by Nabble | Edit this page |