PSF bead shot but X Y and Z dont match

classic Classic list List threaded Threaded
♦
♦
6 messages Options
Matiar Jafari Matiar Jafari
Reply | Threaded
Open this post in threaded view
♦
♦
|

PSF bead shot but X Y and Z dont match

Hey Fellow ListServers,

I purchased a PSF kit from Invitrogen and imaged the beads and when i
loaded the beads to correct the Z blur i measured the XY of the bead
to make sure the diameter matched but it turns out it doesnt.  The
beads are supposedly .17um with +-.005 but whatever angle i measure
the bead at in xy i keep getting .4 um.  Any ideas as to why this is
happening?

Thank You
-------
Matiar Jafari
Fred Mast Fred Mast
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: PSF bead shot but X Y and Z dont match

Is this the diametre of your deconvolved bead? If it is not than you have diffraction to thank for the increased width. The .17 micron beads should be sub-resolution on your microscope and are representative of a point source of light. The resulting "blurred" image that you have captured is the result of the physical properties of light and the optical constraints of your microscope. A good source for further information can be found at the following links:



But also try the Handbook for Biological Confocal Microscopy, James Pawley is the editor, for further information.

Best,

Fred

On 17-Feb-10, at 8:14 PM, Matiar Jafari wrote:

Hey Fellow ListServers,

I purchased a PSF kit from Invitrogen and imaged the beads and when i
loaded the beads to correct the Z blur i measured the XY of the bead
to make sure the diameter matched but it turns out it doesnt.  The
beads are supposedly .17um with +-.005 but whatever angle i measure
the bead at in xy i keep getting .4 um.  Any ideas as to why this is
happening?

Thank You
-------
Matiar Jafari


Fred D. Mast
Department of Cell Biology
Medical Sciences Building Room 5-14
University of Alberta
Edmonton, Alberta, T6G 2H7
Canada

Tel: 1-780-492-7407




Matiar Jafari Matiar Jafari
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: PSF bead shot but X Y and Z dont match

how would i then go about correcting the Z blur for all my other
images. if the XY resolution is just distorted because of the
limitations of optics? sorry im really new to this kind of stuff.
Thank You
-------
Matiar Jafari



On Wed, Feb 17, 2010 at 7:29 PM, Fred Mast <[hidden email]> wrote:

> Is this the diametre of your deconvolved bead? If it is not than you have
> diffraction to thank for the increased width. The .17 micron beads should be
> sub-resolution on your microscope and are representative of a point source
> of light. The resulting "blurred" image that you have captured is the result
> of the physical properties of light and the optical constraints of your
> microscope. A good source for further information can be found at the
> following links:
> http://support.svi.nl/wiki/SubResolution
> http://en.wikipedia.org/wiki/Diffraction
> But also try the Handbook for Biological Confocal Microscopy, James Pawley
> is the editor, for further information.
> Best,
> Fred
> On 17-Feb-10, at 8:14 PM, Matiar Jafari wrote:
>
> Hey Fellow ListServers,
>
> I purchased a PSF kit from Invitrogen and imaged the beads and when i
> loaded the beads to correct the Z blur i measured the XY of the bead
> to make sure the diameter matched but it turns out it doesnt.  The
> beads are supposedly .17um with +-.005 but whatever angle i measure
> the bead at in xy i keep getting .4 um.  Any ideas as to why this is
> happening?
>
> Thank You
> -------
> Matiar Jafari
>
>
> Fred D. Mast
> Department of Cell Biology
> Medical Sciences Building Room 5-14
> University of Alberta
> Edmonton, Alberta, T6G 2H7
> Canada
> Tel: 1-780-492-7407
> [hidden email]
>
>
>
>
Fred Mast Fred Mast
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: PSF bead shot but X Y and Z dont match

Email me off list with the specifics of your setup and I can assist you in getting everything up and running.

On 17-Feb-10, at 8:32 PM, Matiar Jafari wrote:

how would i then go about correcting the Z blur for all my other
images. if the XY resolution is just distorted because of the
limitations of optics? sorry im really new to this kind of stuff.
Thank You
-------
Matiar Jafari



On Wed, Feb 17, 2010 at 7:29 PM, Fred Mast <[hidden email]> wrote:
Is this the diametre of your deconvolved bead? If it is not than you have
diffraction to thank for the increased width. The .17 micron beads should be
sub-resolution on your microscope and are representative of a point source
of light. The resulting "blurred" image that you have captured is the result
of the physical properties of light and the optical constraints of your
microscope. A good source for further information can be found at the
following links:
http://support.svi.nl/wiki/SubResolution
http://en.wikipedia.org/wiki/Diffraction
But also try the Handbook for Biological Confocal Microscopy, James Pawley
is the editor, for further information.
Best,
Fred
On 17-Feb-10, at 8:14 PM, Matiar Jafari wrote:

Hey Fellow ListServers,

I purchased a PSF kit from Invitrogen and imaged the beads and when i
loaded the beads to correct the Z blur i measured the XY of the bead
to make sure the diameter matched but it turns out it doesnt.  The
beads are supposedly .17um with +-.005 but whatever angle i measure
the bead at in xy i keep getting .4 um.  Any ideas as to why this is
happening?

Thank You
-------
Matiar Jafari


Fred D. Mast
Department of Cell Biology
Medical Sciences Building Room 5-14
University of Alberta
Edmonton, Alberta, T6G 2H7
Canada
Tel: 1-780-492-7407
[hidden email]






Fred D. Mast
Department of Cell Biology
Medical Sciences Building Room 5-14
University of Alberta
Edmonton, Alberta, T6G 2H7
Canada

Tel: 1-780-492-7407




Chris Baumann Chris Baumann
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: PSF bead shot but X Y and Z dont match

In reply to this post by Matiar Jafari
Matiar,

What is happening is that you are up against the limit of resolution of the optics in your system.  To correct for this, you would use a technique called deconvolution.  There are a number of resources available on deconvolution but a good place to start will be http://micro.magnet.fsu.edu/primer/digitalimaging/deconvolution/deconvolutionhome.html. 

Good luck,

Chris



 


On 2/17/2010 10:32 PM, Matiar Jafari wrote:
how would i then go about correcting the Z blur for all my other
images. if the XY resolution is just distorted because of the
limitations of optics? sorry im really new to this kind of stuff.
Thank You
-------
Matiar Jafari



On Wed, Feb 17, 2010 at 7:29 PM, Fred Mast [hidden email] wrote:
  
Is this the diametre of your deconvolved bead? If it is not than you have
diffraction to thank for the increased width. The .17 micron beads should be
sub-resolution on your microscope and are representative of a point source
of light. The resulting "blurred" image that you have captured is the result
of the physical properties of light and the optical constraints of your
microscope. A good source for further information can be found at the
following links:
http://support.svi.nl/wiki/SubResolution
http://en.wikipedia.org/wiki/Diffraction
But also try the Handbook for Biological Confocal Microscopy, James Pawley
is the editor, for further information.
Best,
Fred
On 17-Feb-10, at 8:14 PM, Matiar Jafari wrote:

Hey Fellow ListServers,

I purchased a PSF kit from Invitrogen and imaged the beads and when i
loaded the beads to correct the Z blur i measured the XY of the bead
to make sure the diameter matched but it turns out it doesnt.  The
beads are supposedly .17um with +-.005 but whatever angle i measure
the bead at in xy i keep getting .4 um.  Any ideas as to why this is
happening?

Thank You
-------
Matiar Jafari


Fred D. Mast
Department of Cell Biology
Medical Sciences Building Room 5-14
University of Alberta
Edmonton, Alberta, T6G 2H7
Canada
Tel: 1-780-492-7407
[hidden email]




    
Guy Cox-2 Guy Cox-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: PSF bead shot but X Y and Z dont match

In reply to this post by Matiar Jafari
PSF bead shot but X Y and Z dont match
What NA objective are you using?  With an NA 1.4 oil immersion objective you should get a FWHM below 250nm.  If you are getting 400nm then something is optically wrong - probably spherical aberration caused by a refractive index mismatch - things get pretty sensitive at this sort of NA.  With a dry NA 0.75 NA objective you should do better than 450nm but only if the preparation is perfect - my advice is always to use an objective with a correction collar if you want to do serious quality imaging with a dry lens.  (Yes, I know this is not a cheap option.)
 
The other thing to bear in mind is that 170nm isn't really small enough to count as a point object for an NA 1.4 objective - so there will be a convolution of the bead with the psf probably taking the image to larger than the true psf.  I use 100nm or even 60nm beads to check with an oil lens - that way you can in practice ignore the contribution of the object.
 
I can't see much point in getting into deconvolution until your system is performing at its optical limit.  (And even then I wouldn't expect too much).
 
                                                                    Guy
 
Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
     http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
http://www.guycox.net


From: Confocal Microscopy List on behalf of Matiar Jafari
Sent: Thu 10/02/18 2:14 PM
To: [hidden email]
Subject: PSF bead shot but X Y and Z dont match

Hey Fellow ListServers,

I purchased a PSF kit from Invitrogen and imaged the beads and when i
loaded the beads to correct the Z blur i measured the XY of the bead
to make sure the diameter matched but it turns out it doesnt.  The
beads are supposedly .17um with +-.005 but whatever angle i measure
the bead at in xy i keep getting .4 um.  Any ideas as to why this is
happening?

Thank You
-------
Matiar Jafari