|
Matiar Jafari |
|
|
Hey Fellow ListServers,
I purchased a PSF kit from Invitrogen and imaged the beads and when i loaded the beads to correct the Z blur i measured the XY of the bead to make sure the diameter matched but it turns out it doesnt. The beads are supposedly .17um with +-.005 but whatever angle i measure the bead at in xy i keep getting .4 um. Any ideas as to why this is happening? Thank You ------- Matiar Jafari |
|
|
Is this the diametre of your deconvolved bead? If it is not than you have diffraction to thank for the increased width. The .17 micron beads should be sub-resolution on your microscope and are representative of a point source of light. The resulting "blurred" image that you have captured is the result of the physical properties of light and the optical constraints of your microscope. A good source for further information can be found at the following links:
But also try the Handbook for Biological Confocal Microscopy, James Pawley is the editor, for further information. Best, Fred On 17-Feb-10, at 8:14 PM, Matiar Jafari wrote:
Fred D. Mast Department of Cell Biology Medical Sciences Building Room 5-14 University of Alberta Edmonton, Alberta, T6G 2H7 Canada Tel: 1-780-492-7407 |
|
Matiar Jafari |
|
|
how would i then go about correcting the Z blur for all my other
images. if the XY resolution is just distorted because of the limitations of optics? sorry im really new to this kind of stuff. Thank You ------- Matiar Jafari On Wed, Feb 17, 2010 at 7:29 PM, Fred Mast <[hidden email]> wrote: > Is this the diametre of your deconvolved bead? If it is not than you have > diffraction to thank for the increased width. The .17 micron beads should be > sub-resolution on your microscope and are representative of a point source > of light. The resulting "blurred" image that you have captured is the result > of the physical properties of light and the optical constraints of your > microscope. A good source for further information can be found at the > following links: > http://support.svi.nl/wiki/SubResolution > http://en.wikipedia.org/wiki/Diffraction > But also try the Handbook for Biological Confocal Microscopy, James Pawley > is the editor, for further information. > Best, > Fred > On 17-Feb-10, at 8:14 PM, Matiar Jafari wrote: > > Hey Fellow ListServers, > > I purchased a PSF kit from Invitrogen and imaged the beads and when i > loaded the beads to correct the Z blur i measured the XY of the bead > to make sure the diameter matched but it turns out it doesnt. The > beads are supposedly .17um with +-.005 but whatever angle i measure > the bead at in xy i keep getting .4 um. Any ideas as to why this is > happening? > > Thank You > ------- > Matiar Jafari > > > Fred D. Mast > Department of Cell Biology > Medical Sciences Building Room 5-14 > University of Alberta > Edmonton, Alberta, T6G 2H7 > Canada > Tel: 1-780-492-7407 > [hidden email] > > > > |
|
|
Email me off list with the specifics of your setup and I can assist you in getting everything up and running.
On 17-Feb-10, at 8:32 PM, Matiar Jafari wrote:
Fred D. Mast Department of Cell Biology Medical Sciences Building Room 5-14 University of Alberta Edmonton, Alberta, T6G 2H7 Canada Tel: 1-780-492-7407 |
|
Chris Baumann |
|
|
In reply to this post by Matiar Jafari
What is happening is that you are up against the limit of resolution of the optics in your system. To correct for this, you would use a technique called deconvolution. There are a number of resources available on deconvolution but a good place to start will be http://micro.magnet.fsu.edu/primer/digitalimaging/deconvolution/deconvolutionhome.html. Good luck, Chris On 2/17/2010 10:32 PM, Matiar Jafari wrote: how would i then go about correcting the Z blur for all my other images. if the XY resolution is just distorted because of the limitations of optics? sorry im really new to this kind of stuff. Thank You ------- Matiar Jafari On Wed, Feb 17, 2010 at 7:29 PM, Fred Mast [hidden email] wrote:Is this the diametre of your deconvolved bead? If it is not than you have diffraction to thank for the increased width. The .17 micron beads should be sub-resolution on your microscope and are representative of a point source of light. The resulting "blurred" image that you have captured is the result of the physical properties of light and the optical constraints of your microscope. A good source for further information can be found at the following links: http://support.svi.nl/wiki/SubResolution http://en.wikipedia.org/wiki/Diffraction But also try the Handbook for Biological Confocal Microscopy, James Pawley is the editor, for further information. Best, Fred On 17-Feb-10, at 8:14 PM, Matiar Jafari wrote: Hey Fellow ListServers, I purchased a PSF kit from Invitrogen and imaged the beads and when i loaded the beads to correct the Z blur i measured the XY of the bead to make sure the diameter matched but it turns out it doesnt. The beads are supposedly .17um with +-.005 but whatever angle i measure the bead at in xy i keep getting .4 um. Any ideas as to why this is happening? Thank You ------- Matiar Jafari Fred D. Mast Department of Cell Biology Medical Sciences Building Room 5-14 University of Alberta Edmonton, Alberta, T6G 2H7 Canada Tel: 1-780-492-7407 [hidden email] |
|
|
In reply to this post by Matiar Jafari
What NA objective are you using? With an NA 1.4 oil immersion objective you should get a FWHM below 250nm. If you are getting 400nm then something is optically wrong - probably spherical aberration caused by a refractive index mismatch - things get pretty sensitive at this sort of NA. With a dry NA 0.75 NA objective you should do better than 450nm but only if the preparation is perfect - my advice is always to use an objective with a correction collar if you want to do serious quality imaging with a dry lens. (Yes, I know this is not a cheap option.)
The other thing to bear in mind is that 170nm isn't really small enough to count as a point object for an NA 1.4 objective - so there will be a convolution of the bead with the psf probably taking the image to larger than the true psf. I use 100nm or even 60nm beads to check with an oil lens - that way you can in practice ignore the contribution of the object.
I can't see much point in getting into deconvolution until your system is performing at its optical limit. (And even then I wouldn't expect too much).
Guy
Optical Imaging Techniques in Cell Biology
by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net From: Confocal Microscopy List on behalf of Matiar Jafari Sent: Thu 10/02/18 2:14 PM To: [hidden email] Subject: PSF bead shot but X Y and Z dont match Hey Fellow ListServers, |
| Free forum by Nabble | Edit this page |