Login  Register

Phalloidin background on transwell membrane

classic Classic list List threaded Threaded
8 messages Options Options
Embed post
Permalink
Renato A. Mortara Renato A. Mortara
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Phalloidin background on transwell membrane

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear all
 
A colleague of mine is trying to label polarized epithelial cells grown on
transwell dishes.
 
However, fluorescent phalloidin is labeling the dish membrane, rendering it
impossible to image the cells.
 
She tried to block with goat serum, BSA with no luck.
 
Any suggestions ?
 
Many thanks !
 
Renato
 
Renato A. Mortara
Parasitology Division
UNIFESP - Escola Paulista de Medicina
Rua Botucatu, 862, 6th floor
São Paulo, SP
04023-062
Brazil
Phone: 55 11 5579-8306
Fax:     55 11 5571-1095
email: [hidden email]
home page: www.ecb.epm.br/~ramortara
Tobias Baskin Tobias Baskin
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Phalloidin background on transwell membrane

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Try non fluorescent phalloidin??

>*****
>To join, leave or search the confocal microscopy listserv, go to:
>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>*****
>
>Dear all
>
>A colleague of mine is trying to label polarized epithelial cells grown on
>transwell dishes.
>
>However, fluorescent phalloidin is labeling the dish membrane, rendering it
>impossible to image the cells.
>
>She tried to block with goat serum, BSA with no luck.
>
>Any suggestions ?
>
>Many thanks !
>
>Renato
>
>Renato A. Mortara
>Parasitology Division
>UNIFESP - Escola Paulista de Medicina
>Rua Botucatu, 862, 6th floor
>São Paulo, SP
>04023-062
>Brazil
>Phone: 55 11 5579-8306
>Fax:     55 11 5571-1095
>email: [hidden email]
>home page: www.ecb.epm.br/~ramortara


--
       _      ____          __   ____
      /  \   /          / \    /   \ \        Tobias I. Baskin
     /   /  /          /   \   \      \         Biology Department
    /_ /   __      /__ \   \       \__    611 N. Pleasant St.
   /      /          /       \   \       \        University of Massachusetts
  /      /          /         \   \       \    Amherst, MA, 01003
/      / ___   /           \   \__/  \ ____
www.bio.umass.edu/biology/baskin
Voice: 413 - 545 - 1533 Fax: 413 - 545 - 3243
Ryan Schreiner Ryan Schreiner
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Phalloidin background on transwell membrane

In reply to this post by Renato A. Mortara
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

We use polycarbonate transwells for fluorescent studies and do not
have problems with filter background, whereas the many of the clear
filters can have increased background signal, the CF series of
fluorophores coupled to phalloidin from Biotium are particularly good.

On Tue, Jun 14, 2011 at 10:30 AM, Renato Mortara <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Dear all
>
> A colleague of mine is trying to label polarized epithelial cells grown on
> transwell dishes.
>
> However, fluorescent phalloidin is labeling the dish membrane, rendering it
> impossible to image the cells.
>
> She tried to block with goat serum, BSA with no luck.
>
> Any suggestions ?
>
> Many thanks !
>
> Renato
>
> Renato A. Mortara
> Parasitology Division
> UNIFESP - Escola Paulista de Medicina
> Rua Botucatu, 862, 6th floor
> São Paulo, SP
> 04023-062
> Brazil
> Phone: 55 11 5579-8306
> Fax:     55 11 5571-1095
> email: [hidden email]
> home page: www.ecb.epm.br/~ramortara
>
Glen MacDonald-2 Glen MacDonald-2
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Phalloidin background on transwell membrane

In reply to this post by Renato A. Mortara
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

What is the fluorophore?  I found that mitotracker red tends to stick to the Millipore membranes but MT green does not.  Red fluorophores tend to be more lipophilic.  
The Probes ImageiT fx might help block the sticking.

Regards,
Glen

Glen MacDonald
Core for Communication Research
Virginia Merrill Bloedel Hearing Research Center
Box 357923
University of Washington
Seattle, WA 98195-7923
(206) 616-4156

On Tue, 14 Jun 2011, Renato Mortara wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Dear all
>
> A colleague of mine is trying to label polarized epithelial cells grown on
> transwell dishes.
>
> However, fluorescent phalloidin is labeling the dish membrane, rendering it
> impossible to image the cells.
>
> She tried to block with goat serum, BSA with no luck.
>
> Any suggestions ?
>
> Many thanks !
>
> Renato
>
> Renato A. Mortara
> Parasitology Division
> UNIFESP - Escola Paulista de Medicina
> Rua Botucatu, 862, 6th floor
> São Paulo, SP
> 04023-062
> Brazil
> Phone: 55 11 5579-8306
> Fax:     55 11 5571-1095
> email: [hidden email]
> home page: www.ecb.epm.br/~ramortara
>
Cameron, Lisa Cameron, Lisa
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Phalloidin background on transwell membrane

In reply to this post by Renato A. Mortara
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I would recommend looking at the dish membrane without any fluorophore labeling
to check for autofluorescence. I've had a couple people use membranes that are
autofluorescent - mostly in the 488nm range.
So red or far red phalloidin would be better in this case, perhaps.
- Lisa


---------------------------------------
Lisa Cameron, Ph.D.
Director of Confocal and Light Microscopy Core
Dana Farber Cancer Institute
Boston, MA 02215
[hidden email]


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Renato Mortara
Sent: Tuesday, June 14, 2011 10:30 AM
To: [hidden email]
Subject: [CONFOCALMICROSCOPY] Phalloidin background on transwell membrane

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Dear all
 
A colleague of mine is trying to label polarized epithelial cells grown on
transwell dishes.
 
However, fluorescent phalloidin is labeling the dish membrane, rendering it
impossible to image the cells.
 
She tried to block with goat serum, BSA with no luck.
 
Any suggestions ?
 
Many thanks !
 
Renato
 
Renato A. Mortara
Parasitology Division
UNIFESP - Escola Paulista de Medicina
Rua Botucatu, 862, 6th floor
São Paulo, SP
04023-062
Brazil
Phone: 55 11 5579-8306
Fax:     55 11 5571-1095
email: [hidden email]
home page: www.ecb.epm.br/~ramortara


The information in this e-mail is intended only for the person to whom it is
addressed. If you believe this e-mail was sent to you in error and the e-mail
contains patient information, please contact the Partners Compliance HelpLine at
http://www.partners.org/complianceline . If the e-mail was sent to you in error
but does not contain patient information, please contact the sender and properly
dispose of the e-mail.
dlevine dlevine
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Phalloidin background on transwell membrane

In reply to this post by Renato A. Mortara
Or you can try a transwell-like device that does not bind dye or autofluoresce.  CytoVu from SiMPore uses a Silicon membrane that is perfect for fluorescent microscopy.  Check it out at SiMPoreStore.com

Daniel Levine
Daniel Levine
--

Technical Support Specialist
SiMPore Inc
585-286-3384
dlevine@SiMPore.com
www.SiMPoreStore.com
George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Phalloidin background on transwell membrane

In reply to this post by Renato A. Mortara
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

use fluorescent anti-actin


On 6/14/2011 10:30 AM, Renato Mortara wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Dear all
>
> A colleague of mine is trying to label polarized epithelial cells grown on
> transwell dishes.
>
> However, fluorescent phalloidin is labeling the dish membrane, rendering it
> impossible to image the cells.
>
> She tried to block with goat serum, BSA with no luck.
>
> Any suggestions ?
>
> Many thanks !
>
> Renato
>
> Renato A. Mortara
> Parasitology Division
> UNIFESP - Escola Paulista de Medicina
> Rua Botucatu, 862, 6th floor
> São Paulo, SP
> 04023-062
> Brazil
> Phone: 55 11 5579-8306
> Fax:     55 11 5571-1095
> email: [hidden email]
> home page: www.ecb.epm.br/~ramortara
>
>    


--


George McNamara, PhD
Analytical Imaging Core Facility
University of Miami
Julian Smith III Julian Smith III
Reply | Threaded
Open this post in threaded view
| More
Print post
Permalink

Re: Phalloidin background on transwell membrane

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

If you're using Alexas or Cys, check that the transwell membrane doesn't
have a huge positive charge on it.  If so, try an overnight rinse in
buffer with 1% BSA. That trick works to compete charge-bonded dye off of
strongly positive (basic) protein secretion bodies in our worm wholemounts.
Julian

On 6/15/11 9:31 PM, George McNamara wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> use fluorescent anti-actin
>
>
> On 6/14/2011 10:30 AM, Renato Mortara wrote:
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Dear all
>>
>> A colleague of mine is trying to label polarized epithelial cells
>> grown on
>> transwell dishes.
>>
>> However, fluorescent phalloidin is labeling the dish membrane,
>> rendering it
>> impossible to image the cells.
>>
>> She tried to block with goat serum, BSA with no luck.
>>
>> Any suggestions ?
>>
>> Many thanks !
>>
>> Renato
>>
>> Renato A. Mortara
>> Parasitology Division
>> UNIFESP - Escola Paulista de Medicina
>> Rua Botucatu, 862, 6th floor
>> São Paulo, SP
>> 04023-062
>> Brazil
>> Phone: 55 11 5579-8306
>> Fax:     55 11 5571-1095
>> email: [hidden email]
>> home page: www.ecb.epm.br/~ramortara
>>
>
>


--
Julian P.S. Smith III
Director, Winthrop Microscopy Facility
Dept. of Biology
Winthrop University
520 Cherry Rd.
Rock Hill, SC  29733

803-323-2111 x6427 (vox)
803-323-3448 (fax)
803-524-2347 (cell)
Research Website www.birdnest.org/smithj
Personal Website www.rociada-east.net