Zdenek Svindrych-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Listers, not exactly a confocal question, but here it goes: Do you know of a simple, reliable, or even barely useable way to segment transmitted light images of yeast or bacteria, such as this one?: https://drive.google.com/file/d/15ke38DZsADu8ahL9m1WzBEiIxIsBzj8O Preferably in ImageJ, but I'm happy to try other platforms. Thanks! Best, zdenek -- -- Zdenek Svindrych, Ph.D. Research Scientist - Microscopy Imaging Specialist Department of Biochemistry and Cell Biology Geisel School of Medicine at Dartmouth |
|
haesleinhuepf |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Zdenek, I recently gave an online lecture introducing students to basic ImageJ Macro where we process very similar images: https://youtu.be/q4ISEWq6q4E Also check out image.sc for alternate approaches: https://forum.image.sc/t/cell-segmentation-in-phase-contrast/943/14<https://forum.image.sc/t/cell-segmentation-in-phase-contrast/943/14?u=haesleinhuepf> I'm sure there are more sophisticated plugins and tools for processing that kind of data, especially on python side, but maybe that's a starting point. Cheers, Robert -- Robert Haase, Dr. rer. medic. Post-doc, Myers Lab Center for Systems Biology Dresden Max Planck Institute of Molecular Cell Biology and Genetics Pfotenhauerstr. 108 01307 Dresden ________________________________ From: Zdenek Svindrych <[hidden email]> on behalf of Zdenek Svindrych <[hidden email]> Sent: Tuesday, 15 September 2020, 23:35 To: [hidden email] Subject: Phase contrast image segmentation ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Listers, not exactly a confocal question, but here it goes: Do you know of a simple, reliable, or even barely useable way to segment transmitted light images of yeast or bacteria, such as this one?: https://drive.google.com/file/d/15ke38DZsADu8ahL9m1WzBEiIxIsBzj8O Preferably in ImageJ, but I'm happy to try other platforms. Thanks! Best, zdenek -- -- Zdenek Svindrych, Ph.D. Research Scientist - Microscopy Imaging Specialist Department of Biochemistry and Cell Biology Geisel School of Medicine at Dartmouth |
|
Montero Llopis, Paula De La Milagrosa |
|
|
In reply to this post by Zdenek Svindrych-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Zdenek, There are several open source tools you can you to do image segmentation of bacterial cells, in most cases using phase contrast and/or fluorescence, but you may be able to use brightfiled/DIC. In most of these tools you can extract a lot of information and get fancy if needed. In many cases they have great documentation and tutorials to get to set up and, in some cases, there is online help. Here are links to a few of them (not comprehensive). The article link even has a figure that compares some of the tools. https://www.microbej.com/ This one is Image J based, so you may be interested in it https://oufti.org/ Standalone and/or MATLAB based if you wanna get fancy https://www.ncbi.nlm.nih.gov/pmc/articles/PMC7027861/ https://drescherlab.org/data/bacstalk/docs/ This one is mostly for stalk measurements I hope this helps. Bacteria can be tricky to segment! Thank you. Best regards, Paula Paula Montero Llopis, PhD MicRoN Director 77 Avenue Louis Pasteur NRB1032 Office 617 432 0177 Cell 203 927 1468 micron.hms.harvard.edu ________________________________ From: Confocal Microscopy List <[hidden email]> on behalf of Zdenek Svindrych <[hidden email]> Sent: Tuesday, September 15, 2020 5:33 PM To: [hidden email] <[hidden email]> Subject: Phase contrast image segmentation ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Listers, not exactly a confocal question, but here it goes: Do you know of a simple, reliable, or even barely useable way to segment transmitted light images of yeast or bacteria, such as this one?: https://drive.google.com/file/d/15ke38DZsADu8ahL9m1WzBEiIxIsBzj8O Preferably in ImageJ, but I'm happy to try other platforms. Thanks! Best, zdenek -- -- Zdenek Svindrych, Ph.D. Research Scientist - Microscopy Imaging Specialist Department of Biochemistry and Cell Biology Geisel School of Medicine at Dartmouth |
|
Benjamin Smith |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Here is a very quick and dirty ImageJ macro to hopefully lay a foundation for developing a more refined approach (see below). Each step in the macro is commented to explain why that step was done. You also may get a much better answer if you post this same question on the ImageJ forum: https://forum.image.sc/ Note that this macro also makes use of one plugin, Morphological Segmentation, so it requires the MorphoLibJ library. Although, this is a good library to have anyway - lots of good tools in there. The result of the macro should look like this: https://drive.google.com/file/d/1680J9Y8zfTYsDBMLqxiC356Nkf2d40FV Cheers, Ben Smith rename("test"); run("8-bit"); //Convert to gray scale run("Bandpass Filter...", "filter_large=10 filter_small=1 suppress=None tolerance=5 autoscale saturate"); //Remove background mottling setAutoThreshold("Yen"); //Threshold dark border around bacteria setOption("BlackBackground", true); run("Convert to Mask"); run("Skeletonize"); //Reduce borders to single line (collapses background to points) run("Analyze Particles...", "size=50-Infinity show=Masks"); //Remove background objects //Not all of the bacteria borders are fully closed so use a watershed to close them run("Distance Map"); //Create a distance map run("Invert"); run("Morphological Segmentation"); //Use the morphological segmentation watershed, which lets you jump over small ridges wait(1000); selectWindow("Morphological Segmentation"); call("inra.ijpb.plugins.MorphologicalSegmentation.segment", "tolerance=3.0", "calculateDams=true", "connectivity=4"); //Specify minimum ridge height to 3 wait(1000); call("inra.ijpb.plugins.MorphologicalSegmentation.setDisplayFormat", "Watershed lines"); //Set output to watershed dams wait(1000); call("inra.ijpb.plugins.MorphologicalSegmentation.createResultImage"); //Generate the output image selectWindow("Morphological Segmentation"); //Close the Morphological segmentation tool run("Close"); selectWindow("Mask of test-watershed-lines"); run("Erode"); //Solidify borders (needed for particle analyzer) run("Analyze Particles...", "size=50-1000 show=[Count Masks] clear"); //Remove background objects selectWindow("Count Masks of Mask of test-watershed-lines"); //Color code the count mask useing the random LUT reds = newArray(0, 231, 25, 27, 112, 89, 53, 77, 155, 153, 91, 19, 25, 135, 187, 40, 52, 39, 147, 72, 92, 23, 5, 202, 52, 50, 108, 66, 238, 46, 136, 8, 162, 126, 75, 60, 135, 167, 136, 244, 161, 204, 158, 112, 192, 2, 98, 207, 233, 121, 191, 18, 210, 151, 252, 155, 139, 42, 188, 170, 127, 151, 38, 16, 227, 120, 137, 2, 72, 71, 214, 130, 194, 96, 102, 33, 92, 47, 211, 226, 237, 132, 244, 45, 82, 73, 239, 198, 118, 191, 170, 238, 177, 246, 66, 5, 44, 3, 18, 40, 226, 94, 130, 216, 68, 89, 150, 40, 96, 155, 95, 5, 132, 88, 124, 241, 143, 15, 1, 220, 46, 220, 216, 155, 190, 4, 98, 33, 142, 153, 185, 61, 153, 160, 164, 102, 75, 8, 26, 231, 160, 245, 119, 69, 95, 137, 232, 249, 89, 220, 54, 3, 43, 134, 172, 31, 106, 234, 132, 1, 29, 65, 12, 107, 52, 196, 95, 165, 4, 101, 124, 240, 203, 50, 142, 231, 5, 23, 43, 198, 41, 80, 66, 232, 86, 151, 126, 51, 122, 111, 162, 192, 148, 210, 5, 251, 40, 176, 32, 2, 22, 91, 180, 31, 218, 9, 10, 178, 27, 115, 54, 199, 163, 10, 180, 222, 102, 198, 202, 220, 96, 145, 237, 127, 172, 51, 64, 47, 84, 37, 196, 147, 244, 206, 244, 43, 170, 186, 162, 157, 125, 230, 212, 140, 186, 127, 12, 48, 0, 92, 168, 73, 228, 121, 248, 255); greens = newArray(0, 217, 131, 18, 66, 198, 59, 170, 108, 8, 167, 207, 248, 138, 215, 204, 95, 249, 102, 66, 1, 138, 71, 33, 209, 80, 72, 87, 129, 24, 200, 98, 16, 160, 47, 1, 224, 49, 126, 41, 235, 205, 84, 249, 222, 125, 5, 58, 90, 137, 73, 98, 109, 167, 91, 106, 68, 23, 10, 241, 51, 35, 77, 112, 243, 235, 53, 41, 21, 134, 106, 60, 231, 107, 124, 96, 131, 62, 213, 220, 78, 96, 44, 106, 128, 96, 129, 164, 59, 86, 45, 87, 237, 192, 171, 93, 223, 206, 158, 42, 255, 175, 142, 177, 48, 67, 221, 123, 125, 182, 135, 31, 163, 211, 220, 28, 81, 113, 235, 236, 30, 104, 238, 227, 18, 134, 193, 15, 122, 151, 6, 132, 0, 128, 88, 20, 130, 87, 127, 55, 13, 76, 15, 124, 251, 160, 177, 179, 14, 7, 189, 49, 210, 77, 63, 132, 17, 84, 214, 63, 206, 252, 184, 168, 109, 129, 89, 70, 66, 151, 151, 100, 27, 123, 35, 210, 112, 183, 135, 97, 19, 7, 159, 221, 46, 253, 42, 229, 101, 187, 10, 33, 152, 138, 236, 208, 117, 230, 98, 200, 139, 200, 255, 19, 219, 183, 181, 3, 43, 143, 130, 75, 21, 228, 121, 208, 51, 82, 41, 233, 56, 220, 190, 176, 136, 108, 88, 118, 228, 206, 11, 170, 236, 232, 204, 241, 241, 69, 71, 28, 143, 207, 52, 188, 183, 80, 4, 222, 162, 30, 213, 228, 119, 142, 10, 255); blues = newArray(0, 43, 203, 219, 253, 158, 5, 190, 72, 11, 23, 233, 220, 246, 53, 10, 90, 49, 215, 182, 74, 50, 27, 107, 39, 48, 192, 134, 247, 89, 14, 3, 67, 65, 30, 136, 78, 129, 178, 138, 186, 204, 5, 160, 18, 103, 255, 162, 42, 128, 213, 204, 49, 80, 181, 130, 60, 185, 31, 203, 184, 89, 108, 190, 109, 157, 231, 62, 128, 96, 150, 153, 160, 54, 24, 143, 90, 216, 128, 120, 87, 244, 15, 213, 235, 142, 140, 33, 98, 164, 202, 38, 84, 63, 229, 163, 28, 239, 210, 131, 79, 83, 168, 79, 89, 170, 26, 168, 149, 217, 31, 20, 11, 141, 121, 139, 21, 58, 194, 75, 110, 234, 16, 126, 24, 41, 62, 88, 232, 38, 243, 83, 195, 58, 84, 106, 151, 32, 146, 24, 140, 217, 176, 186, 174, 170, 250, 1, 48, 141, 99, 213, 127, 46, 97, 12, 143, 237, 153, 185, 72, 170, 23, 222, 216, 23, 92, 232, 54, 64, 72, 128, 182, 38, 192, 138, 115, 162, 231, 2, 143, 117, 167, 196, 4, 182, 220, 52, 252, 34, 2, 233, 132, 135, 230, 36, 199, 228, 222, 43, 119, 7, 148, 59, 10, 35, 158, 237, 17, 116, 110, 129, 172, 233, 172, 47, 114, 26, 115, 212, 177, 157, 74, 183, 102, 132, 151, 18, 242, 242, 12, 138, 21, 159, 165, 213, 230, 147, 174, 206, 116, 9, 242, 233, 202, 205, 116, 169, 80, 53, 161, 239, 211, 73, 96, 255); setLut(reds, greens, blues); close("\\Others"); On Tue, Sep 15, 2020 at 6:35 PM MODEL, MICHAEL <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I don't know if that's something that interests you, but we have a paper > on measuring bacterial volume in transmitted light. It also enables simple > segmentation of cells by intensity > > Lababidi, Suzanne L., Mariana Pelts, Moumita Moitra, Laura G. Leff, and > Michael A. Model. "Measurement of bacterial volume by > transmission-through-dye imaging." Journal of microbiological methods 87, > no. 3 (2011): 375-377. > > Best wishes > > Mike > > -----Original Message----- > From: Confocal Microscopy List <[hidden email]> On > Behalf Of Montero Llopis, Paula De La Milagrosa > Sent: Tuesday, September 15, 2020 7:48 PM > To: [hidden email] > Subject: Re: Phase contrast image segmentation > > ***** > To join, leave or search the confocal microscopy listserv, go to: > > https://nam03.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7Cmmodel%40KENT.EDU%7C7a77282f61484f09502708d859d1f94b%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637358105664666917&sdata=tC5TzwWbIAizUZNQjGZZJR%2FVAi5kwDvEAdxQAZOK3Vk%3D&reserved=0 > Post images on > https://nam03.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&data=02%7C01%7Cmmodel%40KENT.EDU%7C7a77282f61484f09502708d859d1f94b%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637358105664666917&sdata=zbnQQOp7qcWsIjkyOeUmoPjKroHQ%2BMXkAx8x3n08cTo%3D&reserved=0 > and include the link in your posting. > ***** > > Hi Zdenek, > > There are several open source tools you can you to do image segmentation > of bacterial cells, in most cases using phase contrast and/or fluorescence, > but you may be able to use brightfiled/DIC. In most of these tools you can > extract a lot of information and get fancy if needed. In many cases they > have great documentation and tutorials to get to set up and, in some cases, > there is online help. Here are links to a few of them (not comprehensive). > The article link even has a figure that compares some of the tools. > > > https://nam03.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.microbej.com%2F&data=02%7C01%7Cmmodel%40KENT.EDU%7C7a77282f61484f09502708d859d1f94b%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637358105664666917&sdata=2eWoJXba3AiwTJxe2VmB%2BJlqgORMQqiaAgDGYA3X99o%3D&reserved=0 > This one is Image J based, so you may be interested in it > > https://nam03.safelinks.protection.outlook.com/?url=https%3A%2F%2Foufti.org%2F&data=02%7C01%7Cmmodel%40KENT.EDU%7C7a77282f61484f09502708d859d1f94b%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637358105664666917&sdata=oKUtM7ZWEy8eySPPrYgbEDKaRFj%2BWzaAlbWEeSUh1SQ%3D&reserved=0 > Standalone and/or MATLAB based if you wanna get fancy > > https://nam03.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.ncbi.nlm.nih.gov%2Fpmc%2Farticles%2FPMC7027861%2F&data=02%7C01%7Cmmodel%40KENT.EDU%7C7a77282f61484f09502708d859d1f94b%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637358105664666917&sdata=bW70ycQjjUEMD5EReiWvDmpt%2BAnH%2FO7FDMBD%2F2CgGUE%3D&reserved=0 > > https://nam03.safelinks.protection.outlook.com/?url=https%3A%2F%2Fdrescherlab.org%2Fdata%2Fbacstalk%2Fdocs%2F&data=02%7C01%7Cmmodel%40KENT.EDU%7C7a77282f61484f09502708d859d1f94b%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637358105664666917&sdata=iDEBuaaS8svdRntF8ecjDFAms6XkjdT1YBLa%2BdjKbxs%3D&reserved=0 > This one is mostly for stalk measurements > > I hope this helps. Bacteria can be tricky to segment! > > Thank you. > > Best regards, > > Paula > > > Paula Montero Llopis, PhD > > MicRoN Director > > > 77 Avenue Louis Pasteur > > NRB1032 > > Office 617 432 0177 > > Cell 203 927 1468 > > > micron.hms.harvard.edu > > > ________________________________ > From: Confocal Microscopy List <[hidden email]> on > behalf of Zdenek Svindrych <[hidden email]> > Sent: Tuesday, September 15, 2020 5:33 PM > To: [hidden email] <[hidden email]> > Subject: Phase contrast image segmentation > > ***** > To join, leave or search the confocal microscopy listserv, go to: > > https://nam03.safelinks.protection.outlook.com/?url=http%3A%2F%2Flists.umn.edu%2Fcgi-bin%2Fwa%3FA0%3Dconfocalmicroscopy&data=02%7C01%7Cmmodel%40KENT.EDU%7C7a77282f61484f09502708d859d1f94b%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637358105664676915&sdata=HJeyLcjvVfS%2Br4orvVqPpJxJ5l5qf56AHhWBExgRUqs%3D&reserved=0 > Post images on > https://nam03.safelinks.protection.outlook.com/?url=http%3A%2F%2Fwww.imgur.com%2F&data=02%7C01%7Cmmodel%40KENT.EDU%7C7a77282f61484f09502708d859d1f94b%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637358105664676915&sdata=r5AmZEaca8ntRlOIDzRr%2F8jlBTkZYaQjzivehK9QGf0%3D&reserved=0 > and include the link in your posting. > ***** > > Dear Listers, > > not exactly a confocal question, but here it goes: Do you know of a > simple, reliable, or even barely useable way to segment transmitted light > images of yeast or bacteria, such as this one?: > > https://nam03.safelinks.protection.outlook.com/?url=https%3A%2F%2Fdrive.google.com%2Ffile%2Fd%2F15ke38DZsADu8ahL9m1WzBEiIxIsBzj8O&data=02%7C01%7Cmmodel%40KENT.EDU%7C7a77282f61484f09502708d859d1f94b%7Ce5a06f4a1ec44d018f73e7dd15f26134%7C1%7C0%7C637358105664676915&sdata=50NVBfKheU3VQ8I47WtIV470t6e1Ru8avIFDv7a1VfM%3D&reserved=0 > Preferably in ImageJ, but I'm happy to try other platforms. > > Thanks! > > Best, zdenek > -- > -- > Zdenek Svindrych, Ph.D. > Research Scientist - Microscopy Imaging Specialist Department of > Biochemistry and Cell Biology Geisel School of Medicine at Dartmouth > -- Benjamin E. Smith, Ph. D. Imaging Specialist, Vision Science University of California, Berkeley 195 Life Sciences Addition Berkeley, CA 94720-3200 Tel (510) 642-9712 Fax (510) 643-6791 e-mail: [hidden email] https://vision.berkeley.edu/faculty/core-grants-nei/core-grant-microscopic-imaging/ |
|
Straatman, Kees (Dr.) |