To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi all, Our Leica SP8 has a pair of HyD detectors, and I've noticed recently that we will get a series of saturated lines during a scan at random, intermittent moments. Judging by their appearance, my guess has been that the detectors were picking up cellular data transmissions. Every time I saw them, someone had a cell phone on them, and having them move the cell phone to the far side of the room, or turning the cell phone off, stopped the lines from occurring. I was wondering if anyone else has noticed this with hybrid detectors, and whether this is normal behavior or something we need to have looked into. Thanks, Ben Smith Benjamin E. Smith, Ph.D. Samuel Roberts Noble Microscopy Laboratory Research Scientist II University of Oklahoma Norman, OK 73019 E-mail: [hidden email] Voice 405-325-4391 FAX 405-325-7619 http://www.microscopy.ou.edu/ |
|
Alison J. North |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Ben, We were warned by the Leica guys not to use cellphones anywhere near the scan head on the SP8 - in fact it has a great big picture of a cellphone with a line through it on the scanhead! - so I guess it is a well-known problem, not a special "feature" of your system. Best wishes, Alison On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi all, > Our Leica SP8 has a pair of HyD detectors, and I've noticed recently that we will get a series of saturated lines during a scan at random, intermittent moments. Judging by their appearance, my guess has been that the detectors were picking up cellular data transmissions. Every time I saw them, someone had a cell phone on them, and having them move the cell phone to the far side of the room, or turning the cell phone off, stopped the lines from occurring. > > I was wondering if anyone else has noticed this with hybrid detectors, and whether this is normal behavior or something we need to have looked into. > > Thanks, > Ben Smith > > Benjamin E. Smith, Ph.D. > Samuel Roberts Noble Microscopy Laboratory > Research Scientist II > University of Oklahoma > Norman, OK 73019 > E-mail: [hidden email] > Voice 405-325-4391 > FAX 405-325-7619 > http://www.microscopy.ou.edu/ -- Alison J. North, Ph.D., Research Associate Professor and Senior Director of the Bio-Imaging Resource Center, The Rockefeller University, 1230 York Avenue, New York, NY 10065. Tel: office ++ 212 327 7488 Tel: lab ++ 212 327 7486 Fax: ++ 212 327 7489 |
|
Romin, Yevgeniy/Sloan Kettering Institute |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Ben We have the same problem with one of our sp5 systems that has HyDs on it. Every once in a while you see a saturated line across the image. We have been careful with cell phones around the system, but these lines tend to happen anyway every once in a while. We had them happen just from a door opening or closing on the room. My impression is that while cell phones could be part of the problem, they are probably not the only ones. There could be plenty of equipment or anything else in the room that sometimes interferes. We have these lines very rarely, but I can't think of a scenario where they are completely guaranteed not to show up. That being said, we have two other inverted systems, an sp5 with 3 HyDs and an SP8 with 2 HyDs. We never see any interference lines on these systems, no matter what is going on in the room. My theory is that the HyDs on the inverted scanheads are hidden away towards the rear of the microscope and are perhaps better shielded from interference then those on the upright system, where they are just sticking out in the open. I communicated this theory to Leica, and they said that this could potentially be an explanation, but they haven't yet come up with any idea for potential shielding of them. Hope this helps, Yevgeniy Jul 9, 2013, в 5:24 PM, "Alison North" <[hidden email]> написал(а): > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Ben, > > We were warned by the Leica guys not to use cellphones anywhere near the scan head on the SP8 - in fact it has a great big picture of a cellphone with a line through it on the scanhead! - so I guess it is a well-known problem, not a special "feature" of your system. > > Best wishes, > Alison > > On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hi all, >> Our Leica SP8 has a pair of HyD detectors, and I've noticed recently that we will get a series of saturated lines during a scan at random, intermittent moments. Judging by their appearance, my guess has been that the detectors were picking up cellular data transmissions. Every time I saw them, someone had a cell phone on them, and having them move the cell phone to the far side of the room, or turning the cell phone off, stopped the lines from occurring. >> >> I was wondering if anyone else has noticed this with hybrid detectors, and whether this is normal behavior or something we need to have looked into. >> >> Thanks, >> Ben Smith >> >> Benjamin E. Smith, Ph.D. >> Samuel Roberts Noble Microscopy Laboratory >> Research Scientist II >> University of Oklahoma >> Norman, OK 73019 >> E-mail: [hidden email] >> Voice 405-325-4391 >> FAX 405-325-7619 >> http://www.microscopy.ou.edu/ > > -- > Alison J. North, Ph.D., > Research Associate Professor and > Senior Director of the Bio-Imaging Resource Center, > The Rockefeller University, > 1230 York Avenue, > New York, > NY 10065. > Tel: office ++ 212 327 7488 > Tel: lab ++ 212 327 7486 > Fax: ++ 212 327 7489 > ===================================================================== Please note that this e-mail and any files transmitted from Memorial Sloan-Kettering Cancer Center may be privileged, confidential, and protected from disclosure under applicable law. If the reader of this message is not the intended recipient, or an employee or agent responsible for delivering this message to the intended recipient, you are hereby notified that any reading, dissemination, distribution, copying, or other use of this communication or any of its attachments is strictly prohibited. If you have received this communication in error, please notify the sender immediately by replying to this message and deleting this message, any attachments, and all copies and backups from your computer. |
|
Steve Alexander |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** You could try to shield them with aluminum foil. I tried that to find a interference source on a different system and it worked! Steve Alexander -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Yevgeniy Romin Sent: Tuesday, July 09, 2013 4:36 PM To: [hidden email] Subject: Re: Possible cell phone interference with hybrid detectors ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Ben We have the same problem with one of our sp5 systems that has HyDs on it. Every once in a while you see a saturated line across the image. We have been careful with cell phones around the system, but these lines tend to happen anyway every once in a while. We had them happen just from a door opening or closing on the room. My impression is that while cell phones could be part of the problem, they are probably not the only ones. There could be plenty of equipment or anything else in the room that sometimes interferes. We have these lines very rarely, but I can't think of a scenario where they are completely guaranteed not to show up. That being said, we have two other inverted systems, an sp5 with 3 HyDs and an SP8 with 2 HyDs. We never see any interference lines on these systems, no matter what is going on in the room. My theory is that the HyDs on the inverted scanheads are hidden away towards the rear of the microscope and are perhaps better shielded from interference then those on the upright system, where they are just sticking out in the open. I communicated this theory to Leica, and they said that this could potentially be an explanation, but they haven't yet come up with any idea for potential shielding of them. Hope this helps, Yevgeniy Jul 9, 2013, в 5:24 PM, "Alison North" <[hidden email]> написал(а): > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Ben, > > We were warned by the Leica guys not to use cellphones anywhere near the scan head on the SP8 - in fact it has a great big picture of a cellphone with a line through it on the scanhead! - so I guess it is a well-known problem, not a special "feature" of your system. > > Best wishes, > Alison > > On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hi all, >> Our Leica SP8 has a pair of HyD detectors, and I've noticed recently that we will get a series of saturated lines during a scan at random, intermittent moments. Judging by their appearance, my guess has been that the detectors were picking up cellular data transmissions. Every time I saw them, someone had a cell phone on them, and having them move the cell phone to the far side of the room, or turning the cell phone off, stopped the lines from occurring. >> >> I was wondering if anyone else has noticed this with hybrid detectors, and whether this is normal behavior or something we need to have looked into. >> >> Thanks, >> Ben Smith >> >> Benjamin E. Smith, Ph.D. >> Samuel Roberts Noble Microscopy Laboratory Research Scientist II >> University of Oklahoma Norman, OK 73019 >> E-mail: [hidden email] >> Voice 405-325-4391 >> FAX 405-325-7619 >> http://www.microscopy.ou.edu/ > > -- > Alison J. North, Ph.D., > Research Associate Professor and > Senior Director of the Bio-Imaging Resource Center, The Rockefeller > University, > 1230 York Avenue, > New York, > NY 10065. > Tel: office ++ 212 327 7488 > Tel: lab ++ 212 327 7486 > Fax: ++ 212 327 7489 > ===================================================================== Please note that this e-mail and any files transmitted from Memorial Sloan-Kettering Cancer Center may be privileged, confidential, and protected from disclosure under applicable law. If the reader of this message is not the intended recipient, or an employee or agent responsible for delivering this message to the intended recipient, you are hereby notified that any reading, dissemination, distribution, copying, or other use of this communication or any of its attachments is strictly prohibited. If you have received this communication in error, please notify the sender immediately by replying to this message and deleting this message, any attachments, and all copies and backups from your computer. |
|
Renato Mortara |
|
|
In reply to this post by Smith, Benjamin E.
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Ben We have a similar array of detectors on an SP5: we DO get cell phone (when they work..) interference and Leica got us some 'no cell' stickers/warnings to be placed in the room, right when we had the system installed. Best, Renato Renato Mortara EPM-UNIFESP www.ecb.epm.br/~ramortara 2013/7/9 Smith, Benjamin E. <[hidden email]> > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi all, > Our Leica SP8 has a pair of HyD detectors, and I've noticed recently > that we will get a series of saturated lines during a scan at random, > intermittent moments. Judging by their appearance, my guess has been that > the detectors were picking up cellular data transmissions. Every time I > saw them, someone had a cell phone on them, and having them move the cell > phone to the far side of the room, or turning the cell phone off, stopped > the lines from occurring. > > I was wondering if anyone else has noticed this with hybrid detectors, > and whether this is normal behavior or something we need to have looked > into. > > Thanks, > Ben Smith > > Benjamin E. Smith, Ph.D. > Samuel Roberts Noble Microscopy Laboratory > Research Scientist II > University of Oklahoma > Norman, OK 73019 > E-mail: [hidden email] > Voice 405-325-4391 > FAX 405-325-7619 > http://www.microscopy.ou.edu/ |
|
Rosemary.White |
|
|
In reply to this post by Smith, Benjamin E.
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Yes, we were advised this would be a problem, and have a cell phone bucket outside the door where people have to leave their phones... cheers, Rosemary Dr Rosemary White CSIRO Plant Industry GPO Box 1600 Canberra, ACT 2601 Australia T 61 2 6246 5475 F 61 2 6246 5334 E [hidden email] On 10/07/13 7:13 AM, "Smith, Benjamin E." <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Hi all, > Our Leica SP8 has a pair of HyD detectors, and I've noticed recently >that we will get a series of saturated lines during a scan at random, >intermittent moments. Judging by their appearance, my guess has been >that the detectors were picking up cellular data transmissions. Every >time I saw them, someone had a cell phone on them, and having them move >the cell phone to the far side of the room, or turning the cell phone >off, stopped the lines from occurring. > > I was wondering if anyone else has noticed this with hybrid >detectors, and whether this is normal behavior or something we need to >have looked into. > >Thanks, > Ben Smith > >Benjamin E. Smith, Ph.D. >Samuel Roberts Noble Microscopy Laboratory >Research Scientist II >University of Oklahoma >Norman, OK 73019 >E-mail: [hidden email] >Voice 405-325-4391 >FAX 405-325-7619 >http://www.microscopy.ou.edu/ |
|
Sylvie Le Guyader |
|
|
In reply to this post by Steve Alexander
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi out there Same experience for us. We were getting random interference lines from our bright field detector on a NikonA1. It was not linked to cell phones though but from the cables of another system that stands very nearby. Shielding the cables with alu foil worked a charm. Sylvie On 9 Jul 2013, at 23:50, "Alexander,Steve" <[hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** You could try to shield them with aluminum foil. I tried that to find a interference source on a different system and it worked! Steve Alexander -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Yevgeniy Romin Sent: Tuesday, July 09, 2013 4:36 PM To: [hidden email] Subject: Re: Possible cell phone interference with hybrid detectors ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Ben We have the same problem with one of our sp5 systems that has HyDs on it. Every once in a while you see a saturated line across the image. We have been careful with cell phones around the system, but these lines tend to happen anyway every once in a while. We had them happen just from a door opening or closing on the room. My impression is that while cell phones could be part of the problem, they are probably not the only ones. There could be plenty of equipment or anything else in the room that sometimes interferes. We have these lines very rarely, but I can't think of a scenario where they are completely guaranteed not to show up. That being said, we have two other inverted systems, an sp5 with 3 HyDs and an SP8 with 2 HyDs. We never see any interference lines on these systems, no matter what is going on in the room. My theory is that the HyDs on the inverted scanheads are hidden away towards the rear of the microscope and are perhaps better shielded from interference then those on the upright system, where they are just sticking out in the open. I communicated this theory to Leica, and they said that this could potentially be an explanation, but they haven't yet come up with any idea for potential shielding of them. Hope this helps, Yevgeniy Jul 9, 2013, в 5:24 PM, "Alison North" <[hidden email]> написал(а): > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Ben, > > We were warned by the Leica guys not to use cellphones anywhere near the scan head on the SP8 - in fact it has a great big picture of a cellphone with a line through it on the scanhead! - so I guess it is a well-known problem, not a special "feature" of your system. > > Best wishes, > Alison > > On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hi all, >> Our Leica SP8 has a pair of HyD detectors, and I've noticed recently that we will get a series of saturated lines during a scan at random, intermittent moments. Judging by their appearance, my guess has been that the detectors were picking up cellular data transmissions. Every time I saw them, someone had a cell phone on them, and having them move the cell phone to the far side of the room, or turning the cell phone off, stopped the lines from occurring. >> >> I was wondering if anyone else has noticed this with hybrid detectors, and whether this is normal behavior or something we need to have looked into. >> >> Thanks, >> Ben Smith >> >> Benjamin E. Smith, Ph.D. >> Samuel Roberts Noble Microscopy Laboratory Research Scientist II >> University of Oklahoma Norman, OK 73019 >> E-mail: [hidden email] >> Voice 405-325-4391 >> FAX 405-325-7619 >> http://www.microscopy.ou.edu/ > > -- > Alison J. North, Ph.D., > Research Associate Professor and > Senior Director of the Bio-Imaging Resource Center, The Rockefeller > University, > 1230 York Avenue, > New York, > NY 10065. > Tel: office ++ 212 327 7488 > Tel: lab ++ 212 327 7486 > Fax: ++ 212 327 7489 ===================================================================== Please note that this e-mail and any files transmitted from Memorial Sloan-Kettering Cancer Center may be privileged, confidential, and protected from disclosure under applicable law. If the reader of this message is not the intended recipient, or an employee or agent responsible for delivering this message to the intended recipient, you are hereby notified that any reading, dissemination, distribution, copying, or other use of this communication or any of its attachments is strictly prohibited. If you have received this communication in error, please notify the sender immediately by replying to this message and deleting this message, any attachments, and all copies and backups from your computer. |
|
|
In reply to this post by Renato Mortara
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Lead foil bags, such as the ones we used to use for protecting film from X-rays in airports, wrapped around the PMTs should solve this problem. Don't know, but maybe even the lighter electrostatic protective bags for EZ Pass, circuit boards, etc. would work. ________________________________________________________ Michael Cammer, Assistant Research Scientist Microscopy Core, NYU Langone Medical Center & Skirball Institute of Biomolecular Medicine Cell: (914) 309-3270 Microscopy Lab: (212) 263-7099 Dustin Lab: (212) 263-3208 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Renato Mortara Sent: Tuesday, July 09, 2013 5:55 PM To: [hidden email] Subject: Re: Possible cell phone interference with hybrid detectors ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Ben We have a similar array of detectors on an SP5: we DO get cell phone (when they work..) interference and Leica got us some 'no cell' stickers/warnings to be placed in the room, right when we had the system installed. Best, Renato Renato Mortara EPM-UNIFESP www.ecb.epm.br/~ramortara 2013/7/9 Smith, Benjamin E. <[hidden email]> > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi all, > Our Leica SP8 has a pair of HyD detectors, and I've noticed > recently that we will get a series of saturated lines during a scan at > random, intermittent moments. Judging by their appearance, my guess > has been that the detectors were picking up cellular data > transmissions. Every time I saw them, someone had a cell phone on > them, and having them move the cell phone to the far side of the room, > or turning the cell phone off, stopped the lines from occurring. > > I was wondering if anyone else has noticed this with hybrid > detectors, and whether this is normal behavior or something we need to > have looked into. > > Thanks, > Ben Smith > > Benjamin E. Smith, Ph.D. > Samuel Roberts Noble Microscopy Laboratory Research Scientist II > University of Oklahoma Norman, OK 73019 > E-mail: [hidden email] > Voice 405-325-4391 > FAX 405-325-7619 > http://www.microscopy.ou.edu/ |
|
Peter Gabriel Pitrone |
|
|
In reply to this post by Sylvie Le Guyader
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hey All, I use Aluminum foil to ward off the aliens stealing my thoughts, and of course to ward off WiFi and Cellphone signals... So yeah, that should help with the HyD detectors. ;-) Pete -- Peter Gabriel Pitrone - TechRMS Microscopy/Imaging Specialist Prof. Dr. Pavel Tomancak group Max Planck Institute for Molecular Biology and Genetics Pfotenhauerstr. 108 01307 Dresden "If a straight line fit is required, obtain only two data points." - Anon. On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote: <|> ***** <|> To join, leave or search the confocal microscopy listserv, go to: <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy <|> ***** <|> <|> Hi out there <|> <|> Same experience for us. We were getting random interference lines from <|> our bright field detector on a NikonA1. It was not linked to cell phones <|> though but from the cables of another system that stands very nearby. <|> Shielding the cables with alu foil worked a charm. <|> Sylvie <|> <|> On 9 Jul 2013, at 23:50, "Alexander,Steve" <[hidden email]> <|> wrote: <|> <|> ***** <|> To join, leave or search the confocal microscopy listserv, go to: <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy <|> ***** <|> <|> You could try to shield them with aluminum foil. I tried that to find a <|> interference source on a different system and it worked! <|> <|> Steve Alexander <|> <|> <|> <|> -----Original Message----- <|> From: Confocal Microscopy List [mailto:[hidden email]] <|> On Behalf Of Yevgeniy Romin <|> Sent: Tuesday, July 09, 2013 4:36 PM <|> To: [hidden email] <|> Subject: Re: Possible cell phone interference with hybrid detectors <|> <|> ***** <|> To join, leave or search the confocal microscopy listserv, go to: <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy <|> ***** <|> <|> Hi Ben <|> <|> We have the same problem with one of our sp5 systems that has HyDs on it. <|> Every once in a while you see a saturated line across the image. We have <|> been careful with cell phones around the system, but these lines tend to <|> happen anyway every once in a while. We had them happen just from a door <|> opening or closing on the room. My impression is that while cell phones <|> could be part of the problem, they are probably not the only ones. There <|> could be plenty of equipment or anything else in the room that sometimes <|> interferes. We have these lines very rarely, but I can't think of a <|> scenario where they are completely guaranteed not to show up. <|> <|> That being said, we have two other inverted systems, an sp5 with 3 HyDs <|> and an SP8 with 2 HyDs. We never see any interference lines on these <|> systems, no matter what is going on in the room. My theory is that the <|> HyDs on the inverted scanheads are hidden away towards the rear of the <|> microscope and are perhaps better shielded from interference then those <|> on the upright system, where they are just sticking out in the open. I <|> communicated this theory to Leica, and they said that this could <|> potentially be an explanation, but they haven't yet come up with any idea <|> for potential shielding of them. <|> <|> Hope this helps, <|> Yevgeniy <|> <|> <|> <|> <|> Jul 9, 2013, × 5:24 PM, "Alison North" <[hidden email]> <|> ÎÁÐÉÓÁÌ(Á): <|> <|>> ***** <|>> To join, leave or search the confocal microscopy listserv, go to: <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy <|>> ***** <|>> <|>> Hi Ben, <|>> <|>> We were warned by the Leica guys not to use cellphones anywhere near <|>> the scan head on the SP8 - in fact it has a great big picture of a <|>> cellphone with a line through it on the scanhead! - so I guess it is a <|>> well-known problem, not a special "feature" of your system. <|>> <|>> Best wishes, <|>> Alison <|>> <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote: <|>>> ***** <|>>> To join, leave or search the confocal microscopy listserv, go to: <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy <|>>> ***** <|>>> <|>>> Hi all, <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed <|>>> recently that we will get a series of saturated lines during a scan <|>>> at random, intermittent moments. Judging by their appearance, my <|>>> guess has been that the detectors were picking up cellular data <|>>> transmissions. Every time I saw them, someone had a cell phone on <|>>> them, and having them move the cell phone to the far side of the <|>>> room, or turning the cell phone off, stopped the lines from <|>>> occurring. <|>>> <|>>> I was wondering if anyone else has noticed this with hybrid <|>>> detectors, and whether this is normal behavior or something we need <|>>> to have looked into. <|>>> <|>>> Thanks, <|>>> Ben Smith <|>>> <|>>> Benjamin E. Smith, Ph.D. <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist II <|>>> University of Oklahoma Norman, OK 73019 <|>>> E-mail: [hidden email] <|>>> Voice 405-325-4391 <|>>> FAX 405-325-7619 <|>>> http://www.microscopy.ou.edu/ <|>> <|>> -- <|>> Alison J. North, Ph.D., <|>> Research Associate Professor and <|>> Senior Director of the Bio-Imaging Resource Center, The Rockefeller <|>> University, <|>> 1230 York Avenue, <|>> New York, <|>> NY 10065. <|>> Tel: office ++ 212 327 7488 <|>> Tel: lab ++ 212 327 7486 <|>> Fax: ++ 212 327 7489 <|> <|> <|> ===================================================================== <|> <|> Please note that this e-mail and any files transmitted from <|> Memorial Sloan-Kettering Cancer Center may be privileged, <|> confidential, <|> and protected from disclosure under applicable law. If the reader of <|> this message is not the intended recipient, or an employee or agent <|> responsible for delivering this message to the intended recipient, <|> you are hereby notified that any reading, dissemination, <|> distribution, <|> copying, or other use of this communication or any of its attachments <|> is strictly prohibited. If you have received this communication in <|> error, please notify the sender immediately by replying to this <|> message <|> and deleting this message, any attachments, and all copies and <|> backups <|> from your computer. <|> |
|
Vladimir Zhukarev |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Yep, No miracles, no aliens here. Electronics is a science of electric contacts, as we all know. It takes just one contact with out of range stray capacitance, so that with rest of the electric transmittance chain this will create a sensitive antenna (receiver) for specific radio frequencies. Purely manufacturing issue. Needs clean up from the provider. Cheers Vladimir Vladimir Zhukarev, Ph.D Advanced Microscopy Consulting Life Science and Bio-Medical Applications 'From Image-to Discovery!' [hidden email] 610.350.1510 www.from-image-to-discovery.com --------- Original Message --------- Subject: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)... From: "Peter Gabriel Pitrone" <[hidden email]> Date: 7/11/13 2:35 am To: [hidden email] ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hey All, I use Aluminum foil to ward off the aliens stealing my thoughts, and of course to ward off WiFi and Cellphone signals... So yeah, that should help with the HyD detectors. ;-) Pete -- Peter Gabriel Pitrone - TechRMS Microscopy/Imaging Specialist Prof. Dr. Pavel Tomancak group Max Planck Institute for Molecular Biology and Genetics Pfotenhauerstr. 108 01307 Dresden "If a straight line fit is required, obtain only two data points." - Anon. On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote: <|> ***** <|> To join, leave or search the confocal microscopy listserv, go to: <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy <|> ***** <|> <|> Hi out there <|> <|> Same experience for us. We were getting random interference lines from <|> our bright field detector on a NikonA1. It was not linked to cell phones <|> though but from the cables of another system that stands very nearby. <|> Shielding the cables with alu foil worked a charm. <|> Sylvie <|> <|> On 9 Jul 2013, at 23:50, "Alexander,Steve" <[hidden email]> <|> wrote: <|> <|> ***** <|> To join, leave or search the confocal microscopy listserv, go to: <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy <|> ***** <|> <|> You could try to shield them with aluminum foil. I tried that to find a <|> interference source on a different system and it worked! <|> <|> Steve Alexander <|> <|> <|> <|> -----Original Message----- <|> From: Confocal Microscopy List [mailto:[hidden email]] <|> On Behalf Of Yevgeniy Romin <|> Sent: Tuesday, July 09, 2013 4:36 PM <|> To: [hidden email] <|> Subject: Re: Possible cell phone interference with hybrid detectors <|> <|> ***** <|> To join, leave or search the confocal microscopy listserv, go to: <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy <|> ***** <|> <|> Hi Ben <|> <|> We have the same problem with one of our sp5 systems that has HyDs on it. <|> Every once in a while you see a saturated line across the image. We have <|> been careful with cell phones around the system, but these lines tend to <|> happen anyway every once in a while. We had them happen just from a door <|> opening or closing on the room. My impression is that while cell phones <|> could be part of the problem, they are probably not the only ones. There <|> could be plenty of equipment or anything else in the room that sometimes <|> interferes. We have these lines very rarely, but I can't think of a <|> scenario where they are completely guaranteed not to show up. <|> <|> That being said, we have two other inverted systems, an sp5 with 3 HyDs <|> and an SP8 with 2 HyDs. We never see any interference lines on these <|> systems, no matter what is going on in the room. My theory is that the <|> HyDs on the inverted scanheads are hidden away towards the rear of the <|> microscope and are perhaps better shielded from interference then those <|> on the upright system, where they are just sticking out in the open. I <|> communicated this theory to Leica, and they said that this could <|> potentially be an explanation, but they haven't yet come up with any idea <|> for potential shielding of them. <|> <|> Hope this helps, <|> Yevgeniy <|> <|> <|> <|> <|> Jul 9, 2013, × 5:24 PM, "Alison North" <[hidden email]> <|> ÎÁÐÉÓÁÌ(Á): <|> <|>> ***** <|>> To join, leave or search the confocal microscopy listserv, go to: <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy <|>> ***** <|>> <|>> Hi Ben, <|>> <|>> We were warned by the Leica guys not to use cellphones anywhere near <|>> the scan head on the SP8 - in fact it has a great big picture of a <|>> cellphone with a line through it on the scanhead! - so I guess it is a <|>> well-known problem, not a special "feature" of your system. <|>> <|>> Best wishes, <|>> Alison <|>> <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote: <|>>> ***** <|>>> To join, leave or search the confocal microscopy listserv, go to: <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy <|>>> ***** <|>>> <|>>> Hi all, <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed <|>>> recently that we will get a series of saturated lines during a scan <|>>> at random, intermittent moments. Judging by their appearance, my <|>>> guess has been that the detectors were picking up cellular data <|>>> transmissions. Every time I saw them, someone had a cell phone on <|>>> them, and having them move the cell phone to the far side of the <|>>> room, or turning the cell phone off, stopped the lines from <|>>> occurring. <|>>> <|>>> I was wondering if anyone else has noticed this with hybrid <|>>> detectors, and whether this is normal behavior or something we need <|>>> to have looked into. <|>>> <|>>> Thanks, <|>>> Ben Smith <|>>> <|>>> Benjamin E. Smith, Ph.D. <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist II <|>>> University of Oklahoma Norman, OK 73019 <|>>> E-mail: [hidden email] <|>>> Voice 405-325-4391 <|>>> FAX 405-325-7619 <|>>> http://www.microscopy.ou.edu/ <|>> <|>> -- <|>> Alison J. North, Ph.D., <|>> Research Associate Professor and <|>> Senior Director of the Bio-Imaging Resource Center, The Rockefeller <|>> University, <|>> 1230 York Avenue, <|>> New York, <|>> NY 10065. <|>> Tel: office ++ 212 327 7488 <|>> Tel: lab ++ 212 327 7486 <|>> Fax: ++ 212 327 7489 <|> <|> <|> ===================================================================== <|> <|> Please note that this e-mail and any files transmitted from <|> Memorial Sloan-Kettering Cancer Center may be privileged, <|> confidential, <|> and protected from disclosure under applicable law. If the reader of <|> this message is not the intended recipient, or an employee or agent <|> responsible for delivering this message to the intended recipient, <|> you are hereby notified that any reading, dissemination, <|> distribution, <|> copying, or other use of this communication or any of its attachments <|> is strictly prohibited. If you have received this communication in <|> error, please notify the sender immediately by replying to this <|> message <|> and deleting this message, any attachments, and all copies and <|> backups <|> from your computer. <|> |
|
Tim Feinstein-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I am a little concerned about the use of aluminum as a blocking agent here. Not only is aluminum inert as far as magnets are concerned, but its use to block mind control waves in the radio spectrum is in fact counterproductive, as shown by a research team at MIT in 2005: [...] According to several (apparent) students from MIT who tested several hat designs, there was "a 30 db amplification at 2.6 Ghz and a 20 db amplification at 1.2 Ghz, regardless of the position of the antenna on the cranium." The students note that "the helmets amplify frequency bands that coincide with those allocated to the US government between 1.2 Ghz and 1.4 Ghz. According to the FCC, These bands are supposedly reserved for 'radio location' (ie, GPS), and other communications with satellites."[...] http://www.theregister.co.uk/2005/11/11/tinfoil_hats_as_government_plot/ The original results have since disappeared from the web, suspiciously enough. Though it may coincide with the student's personal web page expiring when he graduated I think we can agree that the episode raises more questions than it answers. Cheers, TF Timothy Feinstein, Ph.D. On Jul 11, 2013, at 10:44 AM, Vladimir Zhukarev <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Yep, > No miracles, no aliens here. Electronics is a science of electric contacts, as we all know. It takes just one contact with out of range stray capacitance, so that with rest of the electric transmittance chain this will create a sensitive antenna (receiver) for specific radio frequencies. Purely manufacturing issue. Needs clean up from the provider. > Cheers > Vladimir > > Vladimir Zhukarev, Ph.D > Advanced Microscopy Consulting > Life Science and Bio-Medical Applications > > 'From Image-to Discovery!' > [hidden email] > 610.350.1510 www.from-image-to-discovery.com > > --------- Original Message --------- Subject: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)... > From: "Peter Gabriel Pitrone" <[hidden email]> > Date: 7/11/13 2:35 am > To: [hidden email] > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hey All, > > I use Aluminum foil to ward off the aliens stealing my thoughts, and of > course to ward off WiFi and Cellphone signals... So yeah, that should > help with the HyD detectors. ;-) > > Pete > > -- > Peter Gabriel Pitrone - TechRMS > Microscopy/Imaging Specialist > Prof. Dr. Pavel Tomancak group > Max Planck Institute for > Molecular Biology and Genetics > Pfotenhauerstr. 108 > 01307 Dresden > > "If a straight line fit is required, obtain only two data points." - Anon. > > > On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote: > <|> ***** > <|> To join, leave or search the confocal microscopy listserv, go to: > <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > <|> ***** > <|> > <|> Hi out there > <|> > <|> Same experience for us. We were getting random interference lines from > <|> our bright field detector on a NikonA1. It was not linked to cell phones > <|> though but from the cables of another system that stands very nearby. > <|> Shielding the cables with alu foil worked a charm. > <|> Sylvie > <|> > <|> On 9 Jul 2013, at 23:50, "Alexander,Steve" <[hidden email]> > <|> wrote: > <|> > <|> ***** > <|> To join, leave or search the confocal microscopy listserv, go to: > <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > <|> ***** > <|> > <|> You could try to shield them with aluminum foil. I tried that to find a > <|> interference source on a different system and it worked! > <|> > <|> Steve Alexander > <|> > <|> > <|> > <|> -----Original Message----- > <|> From: Confocal Microscopy List [mailto:[hidden email]] > <|> On Behalf Of Yevgeniy Romin > <|> Sent: Tuesday, July 09, 2013 4:36 PM > <|> To: [hidden email] > <|> Subject: Re: Possible cell phone interference with hybrid detectors > <|> > <|> ***** > <|> To join, leave or search the confocal microscopy listserv, go to: > <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > <|> ***** > <|> > <|> Hi Ben > <|> > <|> We have the same problem with one of our sp5 systems that has HyDs on > it. > <|> Every once in a while you see a saturated line across the image. We have > <|> been careful with cell phones around the system, but these lines tend to > <|> happen anyway every once in a while. We had them happen just from a door > <|> opening or closing on the room. My impression is that while cell phones > <|> could be part of the problem, they are probably not the only ones. There > <|> could be plenty of equipment or anything else in the room that sometimes > <|> interferes. We have these lines very rarely, but I can't think of a > <|> scenario where they are completely guaranteed not to show up. > <|> > <|> That being said, we have two other inverted systems, an sp5 with 3 HyDs > <|> and an SP8 with 2 HyDs. We never see any interference lines on these > <|> systems, no matter what is going on in the room. My theory is that the > <|> HyDs on the inverted scanheads are hidden away towards the rear of the > <|> microscope and are perhaps better shielded from interference then those > <|> on the upright system, where they are just sticking out in the open. I > <|> communicated this theory to Leica, and they said that this could > <|> potentially be an explanation, but they haven't yet come up with any > idea > <|> for potential shielding of them. > <|> > <|> Hope this helps, > <|> Yevgeniy > <|> > <|> > <|> > <|> > <|> Jul 9, 2013, × 5:24 PM, "Alison North" <[hidden email]> > <|> ÎÁÐÉÓÁÌ(Á): > <|> > <|>> ***** > <|>> To join, leave or search the confocal microscopy listserv, go to: > <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > <|>> ***** > <|>> > <|>> Hi Ben, > <|>> > <|>> We were warned by the Leica guys not to use cellphones anywhere near > <|>> the scan head on the SP8 - in fact it has a great big picture of a > <|>> cellphone with a line through it on the scanhead! - so I guess it is a > <|>> well-known problem, not a special "feature" of your system. > <|>> > <|>> Best wishes, > <|>> Alison > <|>> > <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote: > <|>>> ***** > <|>>> To join, leave or search the confocal microscopy listserv, go to: > <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > <|>>> ***** > <|>>> > <|>>> Hi all, > <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed > <|>>> recently that we will get a series of saturated lines during a scan > <|>>> at random, intermittent moments. Judging by their appearance, my > <|>>> guess has been that the detectors were picking up cellular data > <|>>> transmissions. Every time I saw them, someone had a cell phone on > <|>>> them, and having them move the cell phone to the far side of the > <|>>> room, or turning the cell phone off, stopped the lines from > <|>>> occurring. > <|>>> > <|>>> I was wondering if anyone else has noticed this with hybrid > <|>>> detectors, and whether this is normal behavior or something we need > <|>>> to have looked into. > <|>>> > <|>>> Thanks, > <|>>> Ben Smith > <|>>> > <|>>> Benjamin E. Smith, Ph.D. > <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist II > <|>>> University of Oklahoma Norman, OK 73019 > <|>>> E-mail: [hidden email] > <|>>> Voice 405-325-4391 > <|>>> FAX 405-325-7619 > <|>>> http://www.microscopy.ou.edu/ > <|>> > <|>> -- > <|>> Alison J. North, Ph.D., > <|>> Research Associate Professor and > <|>> Senior Director of the Bio-Imaging Resource Center, The Rockefeller > <|>> University, > <|>> 1230 York Avenue, > <|>> New York, > <|>> NY 10065. > <|>> Tel: office ++ 212 327 7488 > <|>> Tel: lab ++ 212 327 7486 > <|>> Fax: ++ 212 327 7489 > <|> > <|> > <|> > ===================================================================== > <|> > <|> Please note that this e-mail and any files transmitted from > <|> Memorial Sloan-Kettering Cancer Center may be privileged, > <|> confidential, > <|> and protected from disclosure under applicable law. If the reader of > <|> this message is not the intended recipient, or an employee or agent > <|> responsible for delivering this message to the intended recipient, > <|> you are hereby notified that any reading, dissemination, > <|> distribution, > <|> copying, or other use of this communication or any of its > attachments > <|> is strictly prohibited. If you have received this communication in > <|> error, please notify the sender immediately by replying to this > <|> message > <|> and deleting this message, any attachments, and all copies and > <|> backups > <|> from your computer. > <|> |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I'm rather concerned with potentially trapping heat and creating noise for the detectors..hope there will be a Leica answer on this. kitty > Date: Thu, 11 Jul 2013 16:53:38 -0400 > From: [hidden email] > Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)... > To: [hidden email] > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I am a little concerned about the use of aluminum as a blocking agent here. Not only is aluminum inert as far as magnets are concerned, but its use to block mind control waves in the radio spectrum is in fact counterproductive, as shown by a research team at MIT in 2005: > > [...] According to several (apparent) students from MIT who tested several hat designs, there was "a 30 db amplification at 2.6 Ghz and a 20 db amplification at 1.2 Ghz, regardless of the position of the antenna on the cranium." > > The students note that "the helmets amplify frequency bands that coincide with those allocated to the US government between 1.2 Ghz and 1.4 Ghz. According to the FCC, These bands are supposedly reserved for 'radio location' (ie, GPS), and other communications with satellites."[...] > > http://www.theregister.co.uk/2005/11/11/tinfoil_hats_as_government_plot/ > > The original results have since disappeared from the web, suspiciously enough. Though it may coincide with the student's personal web > page expiring when he graduated I think we can agree that the episode raises more questions than it answers. > > Cheers, > > TF > > Timothy Feinstein, Ph.D. > > On Jul 11, 2013, at 10:44 AM, Vladimir Zhukarev <[hidden email]> wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Yep, > > No miracles, no aliens here. Electronics is a science of electric contacts, as we all know. It takes just one contact with out of range stray capacitance, so that with rest of the electric transmittance chain this will create a sensitive antenna (receiver) for specific radio frequencies. Purely manufacturing issue. Needs clean up from the provider. > > Cheers > > Vladimir > > > > Vladimir Zhukarev, Ph.D > > Advanced Microscopy Consulting > > Life Science and Bio-Medical Applications > > > > 'From Image-to Discovery!' > > [hidden email] > > 610.350.1510 www.from-image-to-discovery.com > > > > --------- Original Message --------- Subject: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)... > > From: "Peter Gabriel Pitrone" <[hidden email]> > > Date: 7/11/13 2:35 am > > To: [hidden email] > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Hey All, > > > > I use Aluminum foil to ward off the aliens stealing my thoughts, and of > > course to ward off WiFi and Cellphone signals... So yeah, that should > > help with the HyD detectors. ;-) > > > > Pete > > > > -- > > Peter Gabriel Pitrone - TechRMS > > Microscopy/Imaging Specialist > > Prof. Dr. Pavel Tomancak group > > Max Planck Institute for > > Molecular Biology and Genetics > > Pfotenhauerstr. 108 > > 01307 Dresden > > > > "If a straight line fit is required, obtain only two data points." - Anon. > > > > > > On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote: > > <|> ***** > > <|> To join, leave or search the confocal microscopy listserv, go to: > > <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > <|> ***** > > <|> > > <|> Hi out there > > <|> > > <|> Same experience for us. We were getting random interference lines from > > <|> our bright field detector on a NikonA1. It was not linked to cell phones > > <|> though but from the cables of another system that stands very nearby. > > <|> Shielding the cables with alu foil worked a charm. > > <|> Sylvie > > <|> > > <|> On 9 Jul 2013, at 23:50, "Alexander,Steve" <[hidden email]> > > <|> wrote: > > <|> > > <|> ***** > > <|> To join, leave or search the confocal microscopy listserv, go to: > > <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > <|> ***** > > <|> > > <|> You could try to shield them with aluminum foil. I tried that to find a > > <|> interference source on a different system and it worked! > > <|> > > <|> Steve Alexander > > <|> > > <|> > > <|> > > <|> -----Original Message----- > > <|> From: Confocal Microscopy List [mailto:[hidden email]] > > <|> On Behalf Of Yevgeniy Romin > > <|> Sent: Tuesday, July 09, 2013 4:36 PM > > <|> To: [hidden email] > > <|> Subject: Re: Possible cell phone interference with hybrid detectors > > <|> > > <|> ***** > > <|> To join, leave or search the confocal microscopy listserv, go to: > > <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > <|> ***** > > <|> > > <|> Hi Ben > > <|> > > <|> We have the same problem with one of our sp5 systems that has HyDs on > > it. > > <|> Every once in a while you see a saturated line across the image. We have > > <|> been careful with cell phones around the system, but these lines tend to > > <|> happen anyway every once in a while. We had them happen just from a door > > <|> opening or closing on the room. My impression is that while cell phones > > <|> could be part of the problem, they are probably not the only ones. There > > <|> could be plenty of equipment or anything else in the room that sometimes > > <|> interferes. We have these lines very rarely, but I can't think of a > > <|> scenario where they are completely guaranteed not to show up. > > <|> > > <|> That being said, we have two other inverted systems, an sp5 with 3 HyDs > > <|> and an SP8 with 2 HyDs. We never see any interference lines on these > > <|> systems, no matter what is going on in the room. My theory is that the > > <|> HyDs on the inverted scanheads are hidden away towards the rear of the > > <|> microscope and are perhaps better shielded from interference then those > > <|> on the upright system, where they are just sticking out in the open. I > > <|> communicated this theory to Leica, and they said that this could > > <|> potentially be an explanation, but they haven't yet come up with any > > idea > > <|> for potential shielding of them. > > <|> > > <|> Hope this helps, > > <|> Yevgeniy > > <|> > > <|> > > <|> > > <|> > > <|> Jul 9, 2013, × 5:24 PM, "Alison North" <[hidden email]> > > <|> ÎÁÐÉÓÁÌ(Á): > > <|> > > <|>> ***** > > <|>> To join, leave or search the confocal microscopy listserv, go to: > > <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > <|>> ***** > > <|>> > > <|>> Hi Ben, > > <|>> > > <|>> We were warned by the Leica guys not to use cellphones anywhere near > > <|>> the scan head on the SP8 - in fact it has a great big picture of a > > <|>> cellphone with a line through it on the scanhead! - so I guess it is a > > <|>> well-known problem, not a special "feature" of your system. > > <|>> > > <|>> Best wishes, > > <|>> Alison > > <|>> > > <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote: > > <|>>> ***** > > <|>>> To join, leave or search the confocal microscopy listserv, go to: > > <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > <|>>> ***** > > <|>>> > > <|>>> Hi all, > > <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed > > <|>>> recently that we will get a series of saturated lines during a scan > > <|>>> at random, intermittent moments. Judging by their appearance, my > > <|>>> guess has been that the detectors were picking up cellular data > > <|>>> transmissions. Every time I saw them, someone had a cell phone on > > <|>>> them, and having them move the cell phone to the far side of the > > <|>>> room, or turning the cell phone off, stopped the lines from > > <|>>> occurring. > > <|>>> > > <|>>> I was wondering if anyone else has noticed this with hybrid > > <|>>> detectors, and whether this is normal behavior or something we need > > <|>>> to have looked into. > > <|>>> > > <|>>> Thanks, > > <|>>> Ben Smith > > <|>>> > > <|>>> Benjamin E. Smith, Ph.D. > > <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist II > > <|>>> University of Oklahoma Norman, OK 73019 > > <|>>> E-mail: [hidden email] > > <|>>> Voice 405-325-4391 > > <|>>> FAX 405-325-7619 > > <|>>> http://www.microscopy.ou.edu/ > > <|>> > > <|>> -- > > <|>> Alison J. North, Ph.D., > > <|>> Research Associate Professor and > > <|>> Senior Director of the Bio-Imaging Resource Center, The Rockefeller > > <|>> University, > > <|>> 1230 York Avenue, > > <|>> New York, > > <|>> NY 10065. > > <|>> Tel: office ++ 212 327 7488 > > <|>> Tel: lab ++ 212 327 7486 > > <|>> Fax: ++ 212 327 7489 > > <|> > > <|> > > <|> > > ===================================================================== > > <|> > > <|> Please note that this e-mail and any files transmitted from > > <|> Memorial Sloan-Kettering Cancer Center may be privileged, > > <|> confidential, > > <|> and protected from disclosure under applicable law. If the reader of > > <|> this message is not the intended recipient, or an employee or agent > > <|> responsible for delivering this message to the intended recipient, > > <|> you are hereby notified that any reading, dissemination, > > <|> distribution, > > <|> copying, or other use of this communication or any of its > > attachments > > <|> is strictly prohibited. If you have received this communication in > > <|> error, please notify the sender immediately by replying to this > > <|> message > > <|> and deleting this message, any attachments, and all copies and > > <|> backups > > <|> from your computer. > > <|> |
|
Mike Buchin |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I think there are fine copper screens/mesh materials that would pass heat but possibly block rf. Think Faraday cages. Mike Michael Buchin Stanford Photonics, Inc. Ph: 650-969-5991 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Kitty Lee Sent: Thursday, July 11, 2013 2:06 PM To: [hidden email] Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)... ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I'm rather concerned with potentially trapping heat and creating noise for the detectors..hope there will be a Leica answer on this. kitty > Date: Thu, 11 Jul 2013 16:53:38 -0400 > From: [hidden email] > Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)... > To: [hidden email] > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I am a little concerned about the use of aluminum as a blocking agent here. Not only is aluminum inert as far as magnets are concerned, but its use to block mind control waves in the radio spectrum is in fact counterproductive, as shown by a research team at MIT in 2005: > > [...] According to several (apparent) students from MIT who tested several hat designs, there was "a 30 db amplification at 2.6 Ghz and a 20 db amplification at 1.2 Ghz, regardless of the position of the antenna on the cranium." > > The students note that "the helmets amplify frequency bands that > coincide with those allocated to the US government between 1.2 Ghz and > 1.4 Ghz. According to the FCC, These bands are supposedly reserved for > 'radio location' (ie, GPS), and other communications with > satellites."[...] > > http://www.theregister.co.uk/2005/11/11/tinfoil_hats_as_government_plo > t/ > > The original results have since disappeared from the web, suspiciously > enough. Though it may coincide with the student's personal web page questions than it answers. > > Cheers, > > TF > > Timothy Feinstein, Ph.D. > > On Jul 11, 2013, at 10:44 AM, Vladimir Zhukarev <[hidden email]> wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Yep, > > No miracles, no aliens here. Electronics is a science of electric contacts, as we all know. It takes just one contact with out of range stray capacitance, so that with rest of the electric transmittance chain this will create a sensitive antenna (receiver) for specific radio frequencies. Purely manufacturing issue. Needs clean up from the provider. > > Cheers > > Vladimir > > > > Vladimir Zhukarev, Ph.D > > Advanced Microscopy Consulting > > Life Science and Bio-Medical Applications > > > > 'From Image-to Discovery!' > > [hidden email] > > 610.350.1510 www.from-image-to-discovery.com > > > > --------- Original Message --------- Subject: Tin Foil Hats (Was: Re: > > From: "Peter Gabriel Pitrone" <[hidden email]> > > Date: 7/11/13 2:35 am > > To: [hidden email] > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Hey All, > > > > I use Aluminum foil to ward off the aliens stealing my thoughts, and > > of course to ward off WiFi and Cellphone signals... So yeah, that > > should help with the HyD detectors. ;-) > > > > Pete > > > > -- > > Peter Gabriel Pitrone - TechRMS > > Microscopy/Imaging Specialist > > Prof. Dr. Pavel Tomancak group > > Max Planck Institute for > > Molecular Biology and Genetics > > Pfotenhauerstr. 108 > > 01307 Dresden > > > > "If a straight line fit is required, obtain only two data points." - > > > > > > On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote: > > <|> ***** > > <|> To join, leave or search the confocal microscopy listserv, go to: > > <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > <|> ***** > > <|> > > <|> Hi out there > > <|> > > <|> Same experience for us. We were getting random interference > > lines from <|> our bright field detector on a NikonA1. It was not > > linked to cell phones <|> though but from the cables of another system > > <|> Shielding the cables with alu foil worked a charm. > > <|> Sylvie > > <|> > > <|> On 9 Jul 2013, at 23:50, "Alexander,Steve" > > <[hidden email]> <|> wrote: > > <|> > > <|> ***** > > <|> To join, leave or search the confocal microscopy listserv, go to: > > <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > <|> ***** > > <|> > > <|> You could try to shield them with aluminum foil. I tried that to > > find a <|> interference source on a different system and it worked! > > <|> > > <|> Steve Alexander > > <|> > > <|> > > <|> > > <|> -----Original Message----- > > <|> From: Confocal Microscopy List > > [mailto:[hidden email]] > > <|> On Behalf Of Yevgeniy Romin > > <|> Sent: Tuesday, July 09, 2013 4:36 PM <|> To: > > [hidden email] <|> Subject: Re: Possible cell > > phone interference with hybrid detectors <|> <|> ***** <|> To join, > > leave or search the confocal microscopy listserv, go to: > > <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > <|> ***** > > <|> > > <|> Hi Ben > > <|> > > <|> We have the same problem with one of our sp5 systems that has > > HyDs on it. > > <|> Every once in a while you see a saturated line across the image. > > We have <|> been careful with cell phones around the system, but > > these lines tend to <|> happen anyway every once in a while. We had > > them happen just from a door <|> opening or closing on the room. My > > impression is that while cell phones <|> could be part of the > > problem, they are probably not the only ones. There <|> could be > > plenty of equipment or anything else in the room that sometimes <|> > > interferes. We have these lines very rarely, but I can't think of a <|> > > <|> > > <|> That being said, we have two other inverted systems, an sp5 with > > 3 HyDs <|> and an SP8 with 2 HyDs. We never see any interference > > lines on these <|> systems, no matter what is going on in the room. > > My theory is that the <|> HyDs on the inverted scanheads are hidden > > away towards the rear of the <|> microscope and are perhaps better > > shielded from interference then those <|> on the upright system, > > where they are just sticking out in the open. I <|> communicated > > this theory to Leica, and they said that this could <|> potentially > > be an explanation, but they haven't yet come up with any idea <|> > > for potential shielding of them. > > <|> > > <|> Hope this helps, > > <|> Yevgeniy > > <|> > > <|> > > <|> > > <|> > > <|> Jul 9, 2013, × 5:24 PM, "Alison North" > > <[hidden email]> <|> > > <|> > > <|>> ***** > > <|>> To join, leave or search the confocal microscopy listserv, go to: > > <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > <|>> ***** > > <|>> > > <|>> Hi Ben, > > <|>> > > <|>> We were warned by the Leica guys not to use cellphones anywhere > > near <|>> the scan head on the SP8 - in fact it has a great big > > picture of a <|>> cellphone with a line through it on the scanhead! > > - so I guess it is a <|>> well-known problem, not a special "feature" of > > <|>> > > <|>> Best wishes, > > <|>> Alison > > <|>> > > <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote: > > <|>>> ***** > > <|>>> To join, leave or search the confocal microscopy listserv, go to: > > <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > <|>>> ***** > > <|>>> > > <|>>> Hi all, > > <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed > > <|>>> recently that we will get a series of saturated lines during a > > scan <|>>> at random, intermittent moments. Judging by their > > appearance, my <|>>> guess has been that the detectors were picking > > up cellular data <|>>> transmissions. Every time I saw them, someone > > had a cell phone on <|>>> them, and having them move the cell phone > > to the far side of the <|>>> room, or turning the cell phone off, > > stopped the lines from <|>>> occurring. > > <|>>> > > <|>>> I was wondering if anyone else has noticed this with hybrid > > <|>>> detectors, and whether this is normal behavior or something we > > need <|>>> to have looked into. > > <|>>> > > <|>>> Thanks, > > <|>>> Ben Smith > > <|>>> > > <|>>> Benjamin E. Smith, Ph.D. > > <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist > > II <|>>> University of Oklahoma Norman, OK 73019 <|>>> E-mail: > > [hidden email] <|>>> Voice 405-325-4391 <|>>> FAX > > 405-325-7619 <|>>> http://www.microscopy.ou.edu/ <|>> <|>> -- <|>> > > Alison J. North, Ph.D., <|>> Research Associate Professor and <|>> > > Senior Director of the Bio-Imaging Resource Center, The Rockefeller > > <|>> University, <|>> 1230 York Avenue, <|>> New York, <|>> NY > > 10065. > > <|>> Tel: office ++ 212 327 7488 > > <|>> Tel: lab ++ 212 327 7486 > > <|>> Fax: ++ 212 327 7489 > > <|> > > <|> > > <|> > > ==================================================================== > > = > > <|> > > <|> Please note that this e-mail and any files transmitted from <|> > > Memorial Sloan-Kettering Cancer Center may be privileged, <|> > > confidential, <|> and protected from disclosure under applicable > > law. If the reader of <|> this message is not the intended > > recipient, or an employee or agent <|> responsible for delivering > > this message to the intended recipient, <|> you are hereby notified > > that any reading, dissemination, <|> distribution, <|> copying, or > > other use of this communication or any of its attachments <|> is > > strictly prohibited. If you have received this communication in <|> > > error, please notify the sender immediately by replying to this <|> > > message <|> and deleting this message, any attachments, and all > > copies and <|> backups <|> from your computer. > > <|> |
|
Rosemary.White |
|
|
In reply to this post by Tim Feinstein-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Heh heh, and it would be useful to control the confocal by mind control... On 12/07/13 6:53 AM, "Tim Feinstein" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >I am a little concerned about the use of aluminum as a blocking agent >here. Not only is aluminum inert as far as magnets are concerned, but >its use to block mind control waves in the radio spectrum is in fact >counterproductive, as shown by a research team at MIT in 2005: > >[...] According to several (apparent) students from MIT who tested >several hat designs, there was "a 30 db amplification at 2.6 Ghz and a 20 >db amplification at 1.2 Ghz, regardless of the position of the antenna on >the cranium." > >The students note that "the helmets amplify frequency bands that coincide >with those allocated to the US government between 1.2 Ghz and 1.4 Ghz. >According to the FCC, These bands are supposedly reserved for 'radio >location' (ie, GPS), and other communications with satellites."[...] > >http://www.theregister.co.uk/2005/11/11/tinfoil_hats_as_government_plot/ > >The original results have since disappeared from the web, suspiciously >enough. Though it may coincide with the student's personal web >page expiring when he graduated I think we can agree that the episode >raises more questions than it answers. > >Cheers, > >TF > >Timothy Feinstein, Ph.D. > >On Jul 11, 2013, at 10:44 AM, Vladimir Zhukarev ><[hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Yep, >> No miracles, no aliens here. Electronics is a science of electric >>contacts, as we all know. It takes just one contact with out of range >>stray capacitance, so that with rest of the electric transmittance chain >>this will create a sensitive antenna (receiver) for specific radio >>frequencies. Purely manufacturing issue. Needs clean up from the >>provider. >> Cheers >> Vladimir >> >> Vladimir Zhukarev, Ph.D >> Advanced Microscopy Consulting >> Life Science and Bio-Medical Applications >> >> 'From Image-to Discovery!' >> [hidden email] >> 610.350.1510 www.from-image-to-discovery.com >> >> --------- Original Message --------- Subject: Tin Foil Hats (Was: Re: >>Possible cell phone interference with hybrid detectors)... >> From: "Peter Gabriel Pitrone" <[hidden email]> >> Date: 7/11/13 2:35 am >> To: [hidden email] >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hey All, >> >> I use Aluminum foil to ward off the aliens stealing my thoughts, and of >> course to ward off WiFi and Cellphone signals... So yeah, that should >> help with the HyD detectors. ;-) >> >> Pete >> >> -- >> Peter Gabriel Pitrone - TechRMS >> Microscopy/Imaging Specialist >> Prof. Dr. Pavel Tomancak group >> Max Planck Institute for >> Molecular Biology and Genetics >> Pfotenhauerstr. 108 >> 01307 Dresden >> >> "If a straight line fit is required, obtain only two data points." - >>Anon. >> >> >> On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote: >> <|> ***** >> <|> To join, leave or search the confocal microscopy listserv, go to: >> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> <|> ***** >> <|> >> <|> Hi out there >> <|> >> <|> Same experience for us. We were getting random interference lines >>from >> <|> our bright field detector on a NikonA1. It was not linked to cell >>phones >> <|> though but from the cables of another system that stands very >>nearby. >> <|> Shielding the cables with alu foil worked a charm. >> <|> Sylvie >> <|> >> <|> On 9 Jul 2013, at 23:50, "Alexander,Steve" >><[hidden email]> >> <|> wrote: >> <|> >> <|> ***** >> <|> To join, leave or search the confocal microscopy listserv, go to: >> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> <|> ***** >> <|> >> <|> You could try to shield them with aluminum foil. I tried that to >>find a >> <|> interference source on a different system and it worked! >> <|> >> <|> Steve Alexander >> <|> >> <|> >> <|> >> <|> -----Original Message----- >> <|> From: Confocal Microscopy List >>[mailto:[hidden email]] >> <|> On Behalf Of Yevgeniy Romin >> <|> Sent: Tuesday, July 09, 2013 4:36 PM >> <|> To: [hidden email] >> <|> Subject: Re: Possible cell phone interference with hybrid detectors >> <|> >> <|> ***** >> <|> To join, leave or search the confocal microscopy listserv, go to: >> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> <|> ***** >> <|> >> <|> Hi Ben >> <|> >> <|> We have the same problem with one of our sp5 systems that has HyDs >>on >> it. >> <|> Every once in a while you see a saturated line across the image. We >>have >> <|> been careful with cell phones around the system, but these lines >>tend to >> <|> happen anyway every once in a while. We had them happen just from a >>door >> <|> opening or closing on the room. My impression is that while cell >>phones >> <|> could be part of the problem, they are probably not the only ones. >>There >> <|> could be plenty of equipment or anything else in the room that >>sometimes >> <|> interferes. We have these lines very rarely, but I can't think of a >> <|> scenario where they are completely guaranteed not to show up. >> <|> >> <|> That being said, we have two other inverted systems, an sp5 with 3 >>HyDs >> <|> and an SP8 with 2 HyDs. We never see any interference lines on these >> <|> systems, no matter what is going on in the room. My theory is that >>the >> <|> HyDs on the inverted scanheads are hidden away towards the rear of >>the >> <|> microscope and are perhaps better shielded from interference then >>those >> <|> on the upright system, where they are just sticking out in the >>open. I >> <|> communicated this theory to Leica, and they said that this could >> <|> potentially be an explanation, but they haven't yet come up with any >> idea >> <|> for potential shielding of them. >> <|> >> <|> Hope this helps, >> <|> Yevgeniy >> <|> >> <|> >> <|> >> <|> >> <|> Jul 9, 2013, × 5:24 PM, "Alison North" >><[hidden email]> >> <|> ÎÁÐÉÓÁÌ(Á): >> <|> >> <|>> ***** >> <|>> To join, leave or search the confocal microscopy listserv, go to: >> <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> <|>> ***** >> <|>> >> <|>> Hi Ben, >> <|>> >> <|>> We were warned by the Leica guys not to use cellphones anywhere >>near >> <|>> the scan head on the SP8 - in fact it has a great big picture of a >> <|>> cellphone with a line through it on the scanhead! - so I guess it >>is a >> <|>> well-known problem, not a special "feature" of your system. >> <|>> >> <|>> Best wishes, >> <|>> Alison >> <|>> >> <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote: >> <|>>> ***** >> <|>>> To join, leave or search the confocal microscopy listserv, go to: >> <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> <|>>> ***** >> <|>>> >> <|>>> Hi all, >> <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed >> <|>>> recently that we will get a series of saturated lines during a >>scan >> <|>>> at random, intermittent moments. Judging by their appearance, my >> <|>>> guess has been that the detectors were picking up cellular data >> <|>>> transmissions. Every time I saw them, someone had a cell phone on >> <|>>> them, and having them move the cell phone to the far side of the >> <|>>> room, or turning the cell phone off, stopped the lines from >> <|>>> occurring. >> <|>>> >> <|>>> I was wondering if anyone else has noticed this with hybrid >> <|>>> detectors, and whether this is normal behavior or something we >>need >> <|>>> to have looked into. >> <|>>> >> <|>>> Thanks, >> <|>>> Ben Smith >> <|>>> >> <|>>> Benjamin E. Smith, Ph.D. >> <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist II >> <|>>> University of Oklahoma Norman, OK 73019 >> <|>>> E-mail: [hidden email] >> <|>>> Voice 405-325-4391 >> <|>>> FAX 405-325-7619 >> <|>>> http://www.microscopy.ou.edu/ >> <|>> >> <|>> -- >> <|>> Alison J. North, Ph.D., >> <|>> Research Associate Professor and >> <|>> Senior Director of the Bio-Imaging Resource Center, The Rockefeller >> <|>> University, >> <|>> 1230 York Avenue, >> <|>> New York, >> <|>> NY 10065. >> <|>> Tel: office ++ 212 327 7488 >> <|>> Tel: lab ++ 212 327 7486 >> <|>> Fax: ++ 212 327 7489 >> <|> >> <|> >> <|> >> ===================================================================== >> <|> >> <|> Please note that this e-mail and any files transmitted from >> <|> Memorial Sloan-Kettering Cancer Center may be privileged, >> <|> confidential, >> <|> and protected from disclosure under applicable law. If the reader of >> <|> this message is not the intended recipient, or an employee or agent >> <|> responsible for delivering this message to the intended recipient, >> <|> you are hereby notified that any reading, dissemination, >> <|> distribution, >> <|> copying, or other use of this communication or any of its >> attachments >> <|> is strictly prohibited. If you have received this communication in >> <|> error, please notify the sender immediately by replying to this >> <|> message >> <|> and deleting this message, any attachments, and all copies and >> <|> backups >> <|> from your computer. >> <|> |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Can somebody explain why it is so essential to have cellphones in the confocal room? Guy -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email] Sent: Friday, 12 July 2013 8:27 AM To: [hidden email] Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)... ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Heh heh, and it would be useful to control the confocal by mind control... On 12/07/13 6:53 AM, "Tim Feinstein" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >I am a little concerned about the use of aluminum as a blocking agent >here. Not only is aluminum inert as far as magnets are concerned, but >its use to block mind control waves in the radio spectrum is in fact >counterproductive, as shown by a research team at MIT in 2005: > >[...] According to several (apparent) students from MIT who tested >several hat designs, there was "a 30 db amplification at 2.6 Ghz and a >20 db amplification at 1.2 Ghz, regardless of the position of the >antenna on the cranium." > >The students note that "the helmets amplify frequency bands that >coincide with those allocated to the US government between 1.2 Ghz and 1.4 Ghz. >According to the FCC, These bands are supposedly reserved for 'radio >location' (ie, GPS), and other communications with satellites."[...] > >http://www.theregister.co.uk/2005/11/11/tinfoil_hats_as_government_plot >/ > >The original results have since disappeared from the web, suspiciously >enough. Though it may coincide with the student's personal web page >expiring when he graduated I think we can agree that the episode raises >more questions than it answers. > >Cheers, > >TF > >Timothy Feinstein, Ph.D. > >On Jul 11, 2013, at 10:44 AM, Vladimir Zhukarev ><[hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Yep, >> No miracles, no aliens here. Electronics is a science of electric >>contacts, as we all know. It takes just one contact with out of range >>stray capacitance, so that with rest of the electric transmittance >>chain this will create a sensitive antenna (receiver) for specific >>radio frequencies. Purely manufacturing issue. Needs clean up from the >>provider. >> Cheers >> Vladimir >> >> Vladimir Zhukarev, Ph.D >> Advanced Microscopy Consulting >> Life Science and Bio-Medical Applications >> >> 'From Image-to Discovery!' >> [hidden email] >> 610.350.1510 www.from-image-to-discovery.com >> >> --------- Original Message --------- Subject: Tin Foil Hats (Was: Re: >>Possible cell phone interference with hybrid detectors)... >> From: "Peter Gabriel Pitrone" <[hidden email]> >> Date: 7/11/13 2:35 am >> To: [hidden email] >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hey All, >> >> I use Aluminum foil to ward off the aliens stealing my thoughts, and >> of course to ward off WiFi and Cellphone signals... So yeah, that >> should help with the HyD detectors. ;-) >> >> Pete >> >> -- >> Peter Gabriel Pitrone - TechRMS >> Microscopy/Imaging Specialist >> Prof. Dr. Pavel Tomancak group >> Max Planck Institute for >> Molecular Biology and Genetics >> Pfotenhauerstr. 108 >> 01307 Dresden >> >> "If a straight line fit is required, obtain only two data points." - >>Anon. >> >> >> On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote: >> <|> ***** >> <|> To join, leave or search the confocal microscopy listserv, go to: >> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> <|> ***** >> <|> >> <|> Hi out there >> <|> >> <|> Same experience for us. We were getting random interference lines >>from <|> our bright field detector on a NikonA1. It was not linked to >>cell phones <|> though but from the cables of another system that >>stands very nearby. >> <|> Shielding the cables with alu foil worked a charm. >> <|> Sylvie >> <|> >> <|> On 9 Jul 2013, at 23:50, "Alexander,Steve" >><[hidden email]> >> <|> wrote: >> <|> >> <|> ***** >> <|> To join, leave or search the confocal microscopy listserv, go to: >> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> <|> ***** >> <|> >> <|> You could try to shield them with aluminum foil. I tried that to >>find a <|> interference source on a different system and it worked! >> <|> >> <|> Steve Alexander >> <|> >> <|> >> <|> >> <|> -----Original Message----- >> <|> From: Confocal Microscopy List >>[mailto:[hidden email]] >> <|> On Behalf Of Yevgeniy Romin >> <|> Sent: Tuesday, July 09, 2013 4:36 PM <|> To: >>[hidden email] >> <|> Subject: Re: Possible cell phone interference with hybrid >>detectors <|> <|> ***** <|> To join, leave or search the confocal >>microscopy listserv, go to: >> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> <|> ***** >> <|> >> <|> Hi Ben >> <|> >> <|> We have the same problem with one of our sp5 systems that has >>HyDs on it. >> <|> Every once in a while you see a saturated line across the image. >>We have <|> been careful with cell phones around the system, but >>these lines tend to <|> happen anyway every once in a while. We had >>them happen just from a door <|> opening or closing on the room. My >>impression is that while cell phones <|> could be part of the >>problem, they are probably not the only ones. >>There >> <|> could be plenty of equipment or anything else in the room that >>sometimes <|> interferes. We have these lines very rarely, but I >>can't think of a <|> scenario where they are completely guaranteed >>not to show up. >> <|> >> <|> That being said, we have two other inverted systems, an sp5 with >>3 HyDs <|> and an SP8 with 2 HyDs. We never see any interference >>lines on these <|> systems, no matter what is going on in the room. >>My theory is that the <|> HyDs on the inverted scanheads are hidden >>away towards the rear of the <|> microscope and are perhaps better >>shielded from interference then those <|> on the upright system, >>where they are just sticking out in the open. I <|> communicated this >>theory to Leica, and they said that this could <|> potentially be an >>explanation, but they haven't yet come up with any idea <|> for >>potential shielding of them. >> <|> >> <|> Hope this helps, >> <|> Yevgeniy >> <|> >> <|> >> <|> >> <|> >> <|> Jul 9, 2013, × 5:24 PM, "Alison North" >><[hidden email]> >> <|> ÎÁÐÉÓÁÌ(Á): >> <|> >> <|>> ***** >> <|>> To join, leave or search the confocal microscopy listserv, go to: >> <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> <|>> ***** >> <|>> >> <|>> Hi Ben, >> <|>> >> <|>> We were warned by the Leica guys not to use cellphones anywhere >>near <|>> the scan head on the SP8 - in fact it has a great big >>picture of a <|>> cellphone with a line through it on the scanhead! - >>so I guess it is a <|>> well-known problem, not a special "feature" >>of your system. >> <|>> >> <|>> Best wishes, >> <|>> Alison >> <|>> >> <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote: >> <|>>> ***** >> <|>>> To join, leave or search the confocal microscopy listserv, go to: >> <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> <|>>> ***** >> <|>>> >> <|>>> Hi all, >> <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed >><|>>> recently that we will get a series of saturated lines during a >>scan <|>>> at random, intermittent moments. Judging by their >>appearance, my <|>>> guess has been that the detectors were picking >>up cellular data <|>>> transmissions. Every time I saw them, someone >>had a cell phone on <|>>> them, and having them move the cell phone >>to the far side of the <|>>> room, or turning the cell phone off, >>stopped the lines from <|>>> occurring. >> <|>>> >> <|>>> I was wondering if anyone else has noticed this with hybrid >><|>>> detectors, and whether this is normal behavior or something we >>need <|>>> to have looked into. >> <|>>> >> <|>>> Thanks, >> <|>>> Ben Smith >> <|>>> >> <|>>> Benjamin E. Smith, Ph.D. >> <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist >>II <|>>> University of Oklahoma Norman, OK 73019 <|>>> E-mail: >>[hidden email] <|>>> Voice 405-325-4391 <|>>> FAX >>405-325-7619 <|>>> http://www.microscopy.ou.edu/ <|>> <|>> -- <|>> >>Alison J. North, Ph.D., <|>> Research Associate Professor and <|>> >>Senior Director of the Bio-Imaging Resource Center, The Rockefeller >><|>> University, <|>> 1230 York Avenue, <|>> New York, <|>> NY >>10065. >> <|>> Tel: office ++ 212 327 7488 >> <|>> Tel: lab ++ 212 327 7486 >> <|>> Fax: ++ 212 327 7489 >> <|> >> <|> >> <|> >> ===================================================================== >> <|> >> <|> Please note that this e-mail and any files transmitted from <|> >>Memorial Sloan-Kettering Cancer Center may be privileged, <|> >>confidential, <|> and protected from disclosure under applicable law. >>If the reader of <|> this message is not the intended recipient, or >>an employee or agent <|> responsible for delivering this message to >>the intended recipient, <|> you are hereby notified that any reading, >>dissemination, <|> distribution, <|> copying, or other use of this >>communication or any of its attachments <|> is strictly prohibited. >>If you have received this communication in <|> error, please notify >>the sender immediately by replying to this <|> message <|> and >>deleting this message, any attachments, and all copies and <|> >>backups <|> from your computer. >> <|> |
|
simon walker (BI)-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Just received on mobile whilst in confocal room.. Sent from my mobile On 12 Jul 2013, at 13:12, "Guy Cox" <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Can somebody explain why it is so essential to have cellphones in the confocal room? > > Guy > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email] > Sent: Friday, 12 July 2013 8:27 AM > To: [hidden email] > Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)... > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Heh heh, and it would be useful to control the confocal by mind control... > > On 12/07/13 6:53 AM, "Tim Feinstein" <[hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> I am a little concerned about the use of aluminum as a blocking agent >> here. Not only is aluminum inert as far as magnets are concerned, but >> its use to block mind control waves in the radio spectrum is in fact >> counterproductive, as shown by a research team at MIT in 2005: >> >> [...] According to several (apparent) students from MIT who tested >> several hat designs, there was "a 30 db amplification at 2.6 Ghz and a >> 20 db amplification at 1.2 Ghz, regardless of the position of the >> antenna on the cranium." >> >> The students note that "the helmets amplify frequency bands that >> coincide with those allocated to the US government between 1.2 Ghz and 1.4 Ghz. >> According to the FCC, These bands are supposedly reserved for 'radio >> location' (ie, GPS), and other communications with satellites."[...] >> >> http://www.theregister.co.uk/2005/11/11/tinfoil_hats_as_government_plot >> / >> >> The original results have since disappeared from the web, suspiciously >> enough. Though it may coincide with the student's personal web page >> expiring when he graduated I think we can agree that the episode raises >> more questions than it answers. >> >> Cheers, >> >> TF >> >> Timothy Feinstein, Ph.D. >> >> On Jul 11, 2013, at 10:44 AM, Vladimir Zhukarev >> <[hidden email]> wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Yep, >>> No miracles, no aliens here. Electronics is a science of electric >>> contacts, as we all know. It takes just one contact with out of range >>> stray capacitance, so that with rest of the electric transmittance >>> chain this will create a sensitive antenna (receiver) for specific >>> radio frequencies. Purely manufacturing issue. Needs clean up from the >>> provider. >>> Cheers >>> Vladimir >>> >>> Vladimir Zhukarev, Ph.D >>> Advanced Microscopy Consulting >>> Life Science and Bio-Medical Applications >>> >>> 'From Image-to Discovery!' >>> [hidden email] >>> 610.350.1510 www.from-image-to-discovery.com >>> >>> --------- Original Message --------- Subject: Tin Foil Hats (Was: Re: >>> Possible cell phone interference with hybrid detectors)... >>> From: "Peter Gabriel Pitrone" <[hidden email]> >>> Date: 7/11/13 2:35 am >>> To: [hidden email] >>> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Hey All, >>> >>> I use Aluminum foil to ward off the aliens stealing my thoughts, and >>> of course to ward off WiFi and Cellphone signals... So yeah, that >>> should help with the HyD detectors. ;-) >>> >>> Pete >>> >>> -- >>> Peter Gabriel Pitrone - TechRMS >>> Microscopy/Imaging Specialist >>> Prof. Dr. Pavel Tomancak group >>> Max Planck Institute for >>> Molecular Biology and Genetics >>> Pfotenhauerstr. 108 >>> 01307 Dresden >>> >>> "If a straight line fit is required, obtain only two data points." - >>> Anon. >>> >>> >>> On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote: >>> <|> ***** >>> <|> To join, leave or search the confocal microscopy listserv, go to: >>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> <|> ***** >>> <|> >>> <|> Hi out there >>> <|> >>> <|> Same experience for us. We were getting random interference lines >>> from <|> our bright field detector on a NikonA1. It was not linked to >>> cell phones <|> though but from the cables of another system that >>> stands very nearby. >>> <|> Shielding the cables with alu foil worked a charm. >>> <|> Sylvie >>> <|> >>> <|> On 9 Jul 2013, at 23:50, "Alexander,Steve" >>> <[hidden email]> >>> <|> wrote: >>> <|> >>> <|> ***** >>> <|> To join, leave or search the confocal microscopy listserv, go to: >>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> <|> ***** >>> <|> >>> <|> You could try to shield them with aluminum foil. I tried that to >>> find a <|> interference source on a different system and it worked! >>> <|> >>> <|> Steve Alexander >>> <|> >>> <|> >>> <|> >>> <|> -----Original Message----- >>> <|> From: Confocal Microscopy List >>> [mailto:[hidden email]] >>> <|> On Behalf Of Yevgeniy Romin >>> <|> Sent: Tuesday, July 09, 2013 4:36 PM <|> To: >>> [hidden email] >>> <|> Subject: Re: Possible cell phone interference with hybrid >>> detectors <|> <|> ***** <|> To join, leave or search the confocal >>> microscopy listserv, go to: >>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> <|> ***** >>> <|> >>> <|> Hi Ben >>> <|> >>> <|> We have the same problem with one of our sp5 systems that has >>> HyDs on it. >>> <|> Every once in a while you see a saturated line across the image. >>> We have <|> been careful with cell phones around the system, but >>> these lines tend to <|> happen anyway every once in a while. We had >>> them happen just from a door <|> opening or closing on the room. My >>> impression is that while cell phones <|> could be part of the >>> problem, they are probably not the only ones. >>> There >>> <|> could be plenty of equipment or anything else in the room that >>> sometimes <|> interferes. We have these lines very rarely, but I >>> can't think of a <|> scenario where they are completely guaranteed >>> not to show up. >>> <|> >>> <|> That being said, we have two other inverted systems, an sp5 with >>> 3 HyDs <|> and an SP8 with 2 HyDs. We never see any interference >>> lines on these <|> systems, no matter what is going on in the room. >>> My theory is that the <|> HyDs on the inverted scanheads are hidden >>> away towards the rear of the <|> microscope and are perhaps better >>> shielded from interference then those <|> on the upright system, >>> where they are just sticking out in the open. I <|> communicated this >>> theory to Leica, and they said that this could <|> potentially be an >>> explanation, but they haven't yet come up with any idea <|> for >>> potential shielding of them. >>> <|> >>> <|> Hope this helps, >>> <|> Yevgeniy >>> <|> >>> <|> >>> <|> >>> <|> >>> <|> Jul 9, 2013, × 5:24 PM, "Alison North" >>> <[hidden email]> >>> <|> ÎÁÐÉÓÁÌ(Á): >>> <|> >>> <|>> ***** >>> <|>> To join, leave or search the confocal microscopy listserv, go to: >>> <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> <|>> ***** >>> <|>> >>> <|>> Hi Ben, >>> <|>> >>> <|>> We were warned by the Leica guys not to use cellphones anywhere >>> near <|>> the scan head on the SP8 - in fact it has a great big >>> picture of a <|>> cellphone with a line through it on the scanhead! - >>> so I guess it is a <|>> well-known problem, not a special "feature" >>> of your system. >>> <|>> >>> <|>> Best wishes, >>> <|>> Alison >>> <|>> >>> <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote: >>> <|>>> ***** >>> <|>>> To join, leave or search the confocal microscopy listserv, go to: >>> <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> <|>>> ***** >>> <|>>> >>> <|>>> Hi all, >>> <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed >>> <|>>> recently that we will get a series of saturated lines during a >>> scan <|>>> at random, intermittent moments. Judging by their >>> appearance, my <|>>> guess has been that the detectors were picking >>> up cellular data <|>>> transmissions. Every time I saw them, someone >>> had a cell phone on <|>>> them, and having them move the cell phone >>> to the far side of the <|>>> room, or turning the cell phone off, >>> stopped the lines from <|>>> occurring. >>> <|>>> >>> <|>>> I was wondering if anyone else has noticed this with hybrid >>> <|>>> detectors, and whether this is normal behavior or something we >>> need <|>>> to have looked into. >>> <|>>> >>> <|>>> Thanks, >>> <|>>> Ben Smith >>> <|>>> >>> <|>>> Benjamin E. Smith, Ph.D. >>> <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist >>> II <|>>> University of Oklahoma Norman, OK 73019 <|>>> E-mail: >>> [hidden email] <|>>> Voice 405-325-4391 <|>>> FAX >>> 405-325-7619 <|>>> http://www.microscopy.ou.edu/ <|>> <|>> -- <|>> >>> Alison J. North, Ph.D., <|>> Research Associate Professor and <|>> >>> Senior Director of the Bio-Imaging Resource Center, The Rockefeller >>> <|>> University, <|>> 1230 York Avenue, <|>> New York, <|>> NY >>> 10065. >>> <|>> Tel: office ++ 212 327 7488 >>> <|>> Tel: lab ++ 212 327 7486 >>> <|>> Fax: ++ 212 327 7489 >>> <|> >>> <|> >>> <|> >>> ===================================================================== >>> <|> >>> <|> Please note that this e-mail and any files transmitted from <|> >>> Memorial Sloan-Kettering Cancer Center may be privileged, <|> >>> confidential, <|> and protected from disclosure under applicable law. >>> If the reader of <|> this message is not the intended recipient, or >>> an employee or agent <|> responsible for delivering this message to >>> the intended recipient, <|> you are hereby notified that any reading, >>> dissemination, <|> distribution, <|> copying, or other use of this >>> communication or any of its attachments <|> is strictly prohibited. >>> If you have received this communication in <|> error, please notify >>> the sender immediately by replying to this <|> message <|> and >>> deleting this message, any attachments, and all copies and <|> >>> backups <|> from your computer. >>> <|> The information transmitted in this email is directed only to the addressee. If you received this in error, please contact the sender and delete this email from your system. The contents of this e-mail are the views of the sender and do not necessarily represent the views of the Babraham Institute. Full conditions at: www.babraham.ac.uk<http://www.babraham.ac.uk/email_disclaimer.html> |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** And why? -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of simon walker Sent: Friday, 12 July 2013 10:16 PM To: [hidden email] Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)... ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Just received on mobile whilst in confocal room.. Sent from my mobile On 12 Jul 2013, at 13:12, "Guy Cox" <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Can somebody explain why it is so essential to have cellphones in the confocal room? > > > Guy > > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] On Behalf Of > [hidden email] > Sent: Friday, 12 July 2013 8:27 AM > To: [hidden email] > Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)... > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Heh heh, and it would be useful to control the confocal by mind control... > > On 12/07/13 6:53 AM, "Tim Feinstein" <[hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> I am a little concerned about the use of aluminum as a blocking agent >> here. Not only is aluminum inert as far as magnets are concerned, >> but its use to block mind control waves in the radio spectrum is in >> fact counterproductive, as shown by a research team at MIT in 2005: >> >> [...] According to several (apparent) students from MIT who tested >> several hat designs, there was "a 30 db amplification at 2.6 Ghz and >> a >> 20 db amplification at 1.2 Ghz, regardless of the position of the >> antenna on the cranium." >> >> The students note that "the helmets amplify frequency bands that >> coincide with those allocated to the US government between 1.2 Ghz and 1.4 Ghz. >> According to the FCC, These bands are supposedly reserved for 'radio >> location' (ie, GPS), and other communications with satellites."[...] >> >> http://www.theregister.co.uk/2005/11/11/tinfoil_hats_as_government_pl >> ot >> / >> >> The original results have since disappeared from the web, >> suspiciously enough. Though it may coincide with the student's >> personal web page expiring when he graduated I think we can agree >> that the episode raises more questions than it answers. >> >> Cheers, >> >> TF >> >> Timothy Feinstein, Ph.D. >> >> On Jul 11, 2013, at 10:44 AM, Vladimir Zhukarev >> <[hidden email]> wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Yep, >>> No miracles, no aliens here. Electronics is a science of electric >>> contacts, as we all know. It takes just one contact with out of >>> range stray capacitance, so that with rest of the electric >>> transmittance chain this will create a sensitive antenna (receiver) >>> for specific radio frequencies. Purely manufacturing issue. Needs >>> clean up from the provider. >>> Cheers >>> Vladimir >>> >>> Vladimir Zhukarev, Ph.D >>> Advanced Microscopy Consulting >>> Life Science and Bio-Medical Applications >>> >>> 'From Image-to Discovery!' >>> [hidden email] >>> 610.350.1510 www.from-image-to-discovery.com >>> >>> --------- Original Message --------- Subject: Tin Foil Hats (Was: Re: >>> Possible cell phone interference with hybrid detectors)... >>> From: "Peter Gabriel Pitrone" <[hidden email]> >>> Date: 7/11/13 2:35 am >>> To: [hidden email] >>> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Hey All, >>> >>> I use Aluminum foil to ward off the aliens stealing my thoughts, and >>> of course to ward off WiFi and Cellphone signals... So yeah, that >>> should help with the HyD detectors. ;-) >>> >>> Pete >>> >>> -- >>> Peter Gabriel Pitrone - TechRMS >>> Microscopy/Imaging Specialist >>> Prof. Dr. Pavel Tomancak group >>> Max Planck Institute for >>> Molecular Biology and Genetics >>> Pfotenhauerstr. 108 >>> 01307 Dresden >>> >>> "If a straight line fit is required, obtain only two data points." - >>> Anon. >>> >>> >>> On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote: >>> <|> ***** >>> <|> To join, leave or search the confocal microscopy listserv, go to: >>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> <|> ***** >>> <|> >>> <|> Hi out there >>> <|> >>> <|> Same experience for us. We were getting random interference >>> lines from <|> our bright field detector on a NikonA1. It was not >>> linked to cell phones <|> though but from the cables of another >>> system that stands very nearby. >>> <|> Shielding the cables with alu foil worked a charm. >>> <|> Sylvie >>> <|> >>> <|> On 9 Jul 2013, at 23:50, "Alexander,Steve" >>> <[hidden email]> >>> <|> wrote: >>> <|> >>> <|> ***** >>> <|> To join, leave or search the confocal microscopy listserv, go to: >>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> <|> ***** >>> <|> >>> <|> You could try to shield them with aluminum foil. I tried that to >>> find a <|> interference source on a different system and it worked! >>> <|> >>> <|> Steve Alexander >>> <|> >>> <|> >>> <|> >>> <|> -----Original Message----- >>> <|> From: Confocal Microscopy List >>> [mailto:[hidden email]] >>> <|> On Behalf Of Yevgeniy Romin >>> <|> Sent: Tuesday, July 09, 2013 4:36 PM <|> To: >>> [hidden email] >>> <|> Subject: Re: Possible cell phone interference with hybrid >>> detectors <|> <|> ***** <|> To join, leave or search the confocal >>> microscopy listserv, go to: >>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> <|> ***** >>> <|> >>> <|> Hi Ben >>> <|> >>> <|> We have the same problem with one of our sp5 systems that has >>> HyDs on it. >>> <|> Every once in a while you see a saturated line across the image. >>> We have <|> been careful with cell phones around the system, but >>> these lines tend to <|> happen anyway every once in a while. We had >>> them happen just from a door <|> opening or closing on the room. My >>> impression is that while cell phones <|> could be part of the >>> problem, they are probably not the only ones. >>> There >>> <|> could be plenty of equipment or anything else in the room that >>> sometimes <|> interferes. We have these lines very rarely, but I >>> can't think of a <|> scenario where they are completely guaranteed >>> not to show up. >>> <|> >>> <|> That being said, we have two other inverted systems, an sp5 with >>> 3 HyDs <|> and an SP8 with 2 HyDs. We never see any interference >>> lines on these <|> systems, no matter what is going on in the room. >>> My theory is that the <|> HyDs on the inverted scanheads are hidden >>> away towards the rear of the <|> microscope and are perhaps better >>> shielded from interference then those <|> on the upright system, >>> where they are just sticking out in the open. I <|> communicated >>> this theory to Leica, and they said that this could <|> potentially >>> be an explanation, but they haven't yet come up with any idea <|> >>> for potential shielding of them. >>> <|> >>> <|> Hope this helps, >>> <|> Yevgeniy >>> <|> >>> <|> >>> <|> >>> <|> >>> <|> Jul 9, 2013, × 5:24 PM, "Alison North" >>> <[hidden email]> >>> <|> ÎÁÐÉÓÁÌ(Á): >>> <|> >>> <|>> ***** >>> <|>> To join, leave or search the confocal microscopy listserv, go to: >>> <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> <|>> ***** >>> <|>> >>> <|>> Hi Ben, >>> <|>> >>> <|>> We were warned by the Leica guys not to use cellphones anywhere >>> near <|>> the scan head on the SP8 - in fact it has a great big >>> picture of a <|>> cellphone with a line through it on the scanhead! >>> - so I guess it is a <|>> well-known problem, not a special "feature" >>> of your system. >>> <|>> >>> <|>> Best wishes, >>> <|>> Alison >>> <|>> >>> <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote: >>> <|>>> ***** >>> <|>>> To join, leave or search the confocal microscopy listserv, go to: >>> <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> <|>>> ***** >>> <|>>> >>> <|>>> Hi all, >>> <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed >>> <|>>> recently that we will get a series of saturated lines during a >>> scan <|>>> at random, intermittent moments. Judging by their >>> appearance, my <|>>> guess has been that the detectors were picking >>> up cellular data <|>>> transmissions. Every time I saw them, >>> someone had a cell phone on <|>>> them, and having them move the >>> cell phone to the far side of the <|>>> room, or turning the cell >>> phone off, stopped the lines from <|>>> occurring. >>> <|>>> >>> <|>>> I was wondering if anyone else has noticed this with hybrid >>> <|>>> detectors, and whether this is normal behavior or something we >>> need <|>>> to have looked into. >>> <|>>> >>> <|>>> Thanks, >>> <|>>> Ben Smith >>> <|>>> >>> <|>>> Benjamin E. Smith, Ph.D. >>> <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist >>> II <|>>> University of Oklahoma Norman, OK 73019 <|>>> E-mail: >>> [hidden email] <|>>> Voice 405-325-4391 <|>>> FAX >>> 405-325-7619 <|>>> http://www.microscopy.ou.edu/ <|>> <|>> -- >>> <|>> Alison J. North, Ph.D., <|>> Research Associate Professor and >>> <|>> Senior Director of the Bio-Imaging Resource Center, The >>> Rockefeller <|>> University, <|>> 1230 York Avenue, <|>> New York, >>> <|>> NY 10065. >>> <|>> Tel: office ++ 212 327 7488 >>> <|>> Tel: lab ++ 212 327 7486 >>> <|>> Fax: ++ 212 327 7489 >>> <|> >>> <|> >>> <|> >>> ==================================================================== >>> = >>> <|> >>> <|> Please note that this e-mail and any files transmitted from <|> >>> Memorial Sloan-Kettering Cancer Center may be privileged, <|> >>> confidential, <|> and protected from disclosure under applicable law. >>> If the reader of <|> this message is not the intended recipient, or >>> an employee or agent <|> responsible for delivering this message to >>> the intended recipient, <|> you are hereby notified that any >>> reading, dissemination, <|> distribution, <|> copying, or other >>> use of this communication or any of its attachments <|> is strictly prohibited. >>> If you have received this communication in <|> error, please notify >>> the sender immediately by replying to this <|> message <|> and >>> deleting this message, any attachments, and all copies and <|> >>> backups <|> from your computer. >>> <|> The information transmitted in this email is directed only to the addressee. If you received this in error, please contact the sender and delete this email from your system. The contents of this e-mail are the views of the sender and do not necessarily represent the views of the Babraham Institute. Full conditions at: www.babraham.ac.uk<http://www.babraham.ac.uk/email_disclaimer.html> |
|
MJ Tomaszewski |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I once got locked in a confocal room because the inside door knob fell off. The sad thing is that the room didn't have a working phone yet, nor did it get cell reception. I had to wait for someone to walk by, get their attention, then get them to find someone that had card key access to the room to let me out. This happened twice. Then I started bringing a tool to the room to reattach the door knob. I would imagine that most people have cell phones that play music now. I would often listen to music while on the scopes, not only did it keep me from falling asleep in the dark room late at night, it sometimes helped with the motion sickness scanning for cells. Cell phones are also good for when your PI is trying to track you down and you're trying to convince him that even though you aren't in his lab, you actually *are* doing work. M On Fri, Jul 12, 2013 at 8:26 AM, Guy Cox <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > And why? > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of simon walker > Sent: Friday, 12 July 2013 10:16 PM > To: [hidden email] > Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with > hybrid detectors)... > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Just received on mobile whilst in confocal room.. > > Sent from my mobile > > On 12 Jul 2013, at 13:12, "Guy Cox" <[hidden email]> wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Can somebody explain why it is so essential to have cellphones in the > confocal room? > > > > > > Guy > > > > -----Original Message----- > > From: Confocal Microscopy List > > [mailto:[hidden email]] On Behalf Of > > [hidden email] > > Sent: Friday, 12 July 2013 8:27 AM > > To: [hidden email] > > Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference > with hybrid detectors)... > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Heh heh, and it would be useful to control the confocal by mind > control... > > > > On 12/07/13 6:53 AM, "Tim Feinstein" <[hidden email]> wrote: > > > >> ***** > >> To join, leave or search the confocal microscopy listserv, go to: > >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >> ***** > >> > >> I am a little concerned about the use of aluminum as a blocking agent > >> here. Not only is aluminum inert as far as magnets are concerned, > >> but its use to block mind control waves in the radio spectrum is in > >> fact counterproductive, as shown by a research team at MIT in 2005: > >> > >> [...] According to several (apparent) students from MIT who tested > >> several hat designs, there was "a 30 db amplification at 2.6 Ghz and > >> a > >> 20 db amplification at 1.2 Ghz, regardless of the position of the > >> antenna on the cranium." > >> > >> The students note that "the helmets amplify frequency bands that > >> coincide with those allocated to the US government between 1.2 Ghz and > 1.4 Ghz. > >> According to the FCC, These bands are supposedly reserved for 'radio > >> location' (ie, GPS), and other communications with satellites."[...] > >> > >> http://www.theregister.co.uk/2005/11/11/tinfoil_hats_as_government_pl > >> ot > >> / > >> > >> The original results have since disappeared from the web, > >> suspiciously enough. Though it may coincide with the student's > >> personal web page expiring when he graduated I think we can agree > >> that the episode raises more questions than it answers. > >> > >> Cheers, > >> > >> TF > >> > >> Timothy Feinstein, Ph.D. > >> > >> On Jul 11, 2013, at 10:44 AM, Vladimir Zhukarev > >> <[hidden email]> wrote: > >> > >>> ***** > >>> To join, leave or search the confocal microscopy listserv, go to: > >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >>> ***** > >>> > >>> Yep, > >>> No miracles, no aliens here. Electronics is a science of electric > >>> contacts, as we all know. It takes just one contact with out of > >>> range stray capacitance, so that with rest of the electric > >>> transmittance chain this will create a sensitive antenna (receiver) > >>> for specific radio frequencies. Purely manufacturing issue. Needs > >>> clean up from the provider. > >>> Cheers > >>> Vladimir > >>> > >>> Vladimir Zhukarev, Ph.D > >>> Advanced Microscopy Consulting > >>> Life Science and Bio-Medical Applications > >>> > >>> 'From Image-to Discovery!' > >>> [hidden email] > >>> 610.350.1510 www.from-image-to-discovery.com > >>> > >>> --------- Original Message --------- Subject: Tin Foil Hats (Was: Re: > >>> Possible cell phone interference with hybrid detectors)... > >>> From: "Peter Gabriel Pitrone" <[hidden email]> > >>> Date: 7/11/13 2:35 am > >>> To: [hidden email] > >>> > >>> ***** > >>> To join, leave or search the confocal microscopy listserv, go to: > >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >>> ***** > >>> > >>> Hey All, > >>> > >>> I use Aluminum foil to ward off the aliens stealing my thoughts, and > >>> of course to ward off WiFi and Cellphone signals... So yeah, that > >>> should help with the HyD detectors. ;-) > >>> > >>> Pete > >>> > >>> -- > >>> Peter Gabriel Pitrone - TechRMS > >>> Microscopy/Imaging Specialist > >>> Prof. Dr. Pavel Tomancak group > >>> Max Planck Institute for > >>> Molecular Biology and Genetics > >>> Pfotenhauerstr. 108 > >>> 01307 Dresden > >>> > >>> "If a straight line fit is required, obtain only two data points." - > >>> Anon. > >>> > >>> > >>> On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote: > >>> <|> ***** > >>> <|> To join, leave or search the confocal microscopy listserv, go to: > >>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >>> <|> ***** > >>> <|> > >>> <|> Hi out there > >>> <|> > >>> <|> Same experience for us. We were getting random interference > >>> lines from <|> our bright field detector on a NikonA1. It was not > >>> linked to cell phones <|> though but from the cables of another > >>> system that stands very nearby. > >>> <|> Shielding the cables with alu foil worked a charm. > >>> <|> Sylvie > >>> <|> > >>> <|> On 9 Jul 2013, at 23:50, "Alexander,Steve" > >>> <[hidden email]> > >>> <|> wrote: > >>> <|> > >>> <|> ***** > >>> <|> To join, leave or search the confocal microscopy listserv, go to: > >>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >>> <|> ***** > >>> <|> > >>> <|> You could try to shield them with aluminum foil. I tried that to > >>> find a <|> interference source on a different system and it worked! > >>> <|> > >>> <|> Steve Alexander > >>> <|> > >>> <|> > >>> <|> > >>> <|> -----Original Message----- > >>> <|> From: Confocal Microscopy List > >>> [mailto:[hidden email]] > >>> <|> On Behalf Of Yevgeniy Romin > >>> <|> Sent: Tuesday, July 09, 2013 4:36 PM <|> To: > >>> [hidden email] > >>> <|> Subject: Re: Possible cell phone interference with hybrid > >>> detectors <|> <|> ***** <|> To join, leave or search the confocal > >>> microscopy listserv, go to: > >>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >>> <|> ***** > >>> <|> > >>> <|> Hi Ben > >>> <|> > >>> <|> We have the same problem with one of our sp5 systems that has > >>> HyDs on it. > >>> <|> Every once in a while you see a saturated line across the image. > >>> We have <|> been careful with cell phones around the system, but > >>> these lines tend to <|> happen anyway every once in a while. We had > >>> them happen just from a door <|> opening or closing on the room. My > >>> impression is that while cell phones <|> could be part of the > >>> problem, they are probably not the only ones. > >>> There > >>> <|> could be plenty of equipment or anything else in the room that > >>> sometimes <|> interferes. We have these lines very rarely, but I > >>> can't think of a <|> scenario where they are completely guaranteed > >>> not to show up. > >>> <|> > >>> <|> That being said, we have two other inverted systems, an sp5 with > >>> 3 HyDs <|> and an SP8 with 2 HyDs. We never see any interference > >>> lines on these <|> systems, no matter what is going on in the room. > >>> My theory is that the <|> HyDs on the inverted scanheads are hidden > >>> away towards the rear of the <|> microscope and are perhaps better > >>> shielded from interference then those <|> on the upright system, > >>> where they are just sticking out in the open. I <|> communicated > >>> this theory to Leica, and they said that this could <|> potentially > >>> be an explanation, but they haven't yet come up with any idea <|> > >>> for potential shielding of them. > >>> <|> > >>> <|> Hope this helps, > >>> <|> Yevgeniy > >>> <|> > >>> <|> > >>> <|> > >>> <|> > >>> <|> Jul 9, 2013, × 5:24 PM, "Alison North" > >>> <[hidden email]> > >>> <|> ÎÁÐÉÓÁÌ(Á): > >>> <|> > >>> <|>> ***** > >>> <|>> To join, leave or search the confocal microscopy listserv, go to: > >>> <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >>> <|>> ***** > >>> <|>> > >>> <|>> Hi Ben, > >>> <|>> > >>> <|>> We were warned by the Leica guys not to use cellphones anywhere > >>> near <|>> the scan head on the SP8 - in fact it has a great big > >>> picture of a <|>> cellphone with a line through it on the scanhead! > >>> - so I guess it is a <|>> well-known problem, not a special "feature" > >>> of your system. > >>> <|>> > >>> <|>> Best wishes, > >>> <|>> Alison > >>> <|>> > >>> <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote: > >>> <|>>> ***** > >>> <|>>> To join, leave or search the confocal microscopy listserv, go to: > >>> <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >>> <|>>> ***** > >>> <|>>> > >>> <|>>> Hi all, > >>> <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed > >>> <|>>> recently that we will get a series of saturated lines during a > >>> scan <|>>> at random, intermittent moments. Judging by their > >>> appearance, my <|>>> guess has been that the detectors were picking > >>> up cellular data <|>>> transmissions. Every time I saw them, > >>> someone had a cell phone on <|>>> them, and having them move the > >>> cell phone to the far side of the <|>>> room, or turning the cell > >>> phone off, stopped the lines from <|>>> occurring. > >>> <|>>> > >>> <|>>> I was wondering if anyone else has noticed this with hybrid > >>> <|>>> detectors, and whether this is normal behavior or something we > >>> need <|>>> to have looked into. > >>> <|>>> > >>> <|>>> Thanks, > >>> <|>>> Ben Smith > >>> <|>>> > >>> <|>>> Benjamin E. Smith, Ph.D. > >>> <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist > >>> II <|>>> University of Oklahoma Norman, OK 73019 <|>>> E-mail: > >>> [hidden email] <|>>> Voice 405-325-4391 <|>>> FAX > >>> 405-325-7619 <|>>> http://www.microscopy.ou.edu/ <|>> <|>> -- > >>> <|>> Alison J. North, Ph.D., <|>> Research Associate Professor and > >>> <|>> Senior Director of the Bio-Imaging Resource Center, The > >>> Rockefeller <|>> University, <|>> 1230 York Avenue, <|>> New York, > >>> <|>> NY 10065. > >>> <|>> Tel: office ++ 212 327 7488 > >>> <|>> Tel: lab ++ 212 327 7486 > >>> <|>> Fax: ++ 212 327 7489 > >>> <|> > >>> <|> > >>> <|> > >>> ==================================================================== > >>> = > >>> <|> > >>> <|> Please note that this e-mail and any files transmitted from <|> > >>> Memorial Sloan-Kettering Cancer Center may be privileged, <|> > >>> confidential, <|> and protected from disclosure under applicable law. > >>> If the reader of <|> this message is not the intended recipient, or > >>> an employee or agent <|> responsible for delivering this message to > >>> the intended recipient, <|> you are hereby notified that any > >>> reading, dissemination, <|> distribution, <|> copying, or other > >>> use of this communication or any of its attachments <|> is strictly > prohibited. > >>> If you have received this communication in <|> error, please notify > >>> the sender immediately by replying to this <|> message <|> and > >>> deleting this message, any attachments, and all copies and <|> > >>> backups <|> from your computer. > >>> <|> > The Babraham Institute, Babraham Research Campus, Cambridge CB22 3AT > Registered Charity No. 1053902. > The information transmitted in this email is directed only to the > addressee. If you received this in error, please contact the sender and > delete this email from your system. The contents of this e-mail are the > views of the sender and do not necessarily represent the views of the > Babraham Institute. Full conditions at: www.babraham.ac.uk< > http://www.babraham.ac.uk/email_disclaimer.html> > |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I have a related question. How is WiFi different from cell phones? In our institution, there is strong WiFi signal everywhere, including our confocal rooms. Would that cause similar interference?? Also, we have MDs on call using the confocals. They are obliged to carry cell phones all the time. We do ask them to not to make calls in the room, but these things are constantly receiving and sending data. Sho Fujisawa Molecular Cytology Core Facility Memorial Sloan-Kettering Cancer Center 415 East 68th St. ZRC-1840 New York, NY 10065 (646) 888-2186, x2186 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of MJ Tomaszewski Sent: Friday, July 12, 2013 8:41 AM To: [hidden email] Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)... ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I once got locked in a confocal room because the inside door knob fell off. The sad thing is that the room didn't have a working phone yet, nor did it get cell reception. I had to wait for someone to walk by, get their attention, then get them to find someone that had card key access to the room to let me out. This happened twice. Then I started bringing a tool to the room to reattach the door knob. I would imagine that most people have cell phones that play music now. I would often listen to music while on the scopes, not only did it keep me from falling asleep in the dark room late at night, it sometimes helped with the motion sickness scanning for cells. Cell phones are also good for when your PI is trying to track you down and you're trying to convince him that even though you aren't in his lab, you actually *are* doing work. M ===================================================================== Please note that this e-mail and any files transmitted from Memorial Sloan-Kettering Cancer Center may be privileged, confidential, and protected from disclosure under applicable law. If the reader of this message is not the intended recipient, or an employee or agent responsible for delivering this message to the intended recipient, you are hereby notified that any reading, dissemination, distribution, copying, or other use of this communication or any of its attachments is strictly prohibited. If you have received this communication in error, please notify the sender immediately by replying to this message and deleting this message, any attachments, and all copies and backups from your computer. |
|
Chris Tully |
|
|
In reply to this post by MJ Tomaszewski
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I will strongly second the idea of a Faraday cage is cell phone signals are causing that much interference. I would even push back against the manufacturer to demand that they retro fit the scope with the necessary shielding. They probably won't but it never hurts to try. I say this because I remember back in the 90's seeing sign all over hospitals to turn off personal cell phones because they interfered with patient care equipment. More recently though those signs have disappeared as equipment has been upgraded or replaced. Back to the Faraday cage idea though: if it is effective enough to shield the experiments looking at electrical signals of mouse muscles then it should shield a scope from most external interference. Just make sure that it fully encloses the components that are picking up the interfering signals. And ground it! Chris Tully, M.S., Image Analysis Expert t 240.475.9753 f 419.831.0527 | [hidden email] Sent from my iPhone please excuse typos. On Jul 12, 2013, at 8:42 AM, MJ Tomaszewski <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I once got locked in a confocal room because the inside door knob fell off. > The sad thing is that the room didn't have a working phone yet, nor did it > get cell reception. I had to wait for someone to walk by, get their > attention, then get them to find someone that had card key access to the > room to let me out. This happened twice. Then I started bringing a tool to > the room to reattach the door knob. > > I would imagine that most people have cell phones that play music now. I > would often listen to music while on the scopes, not only did it keep me > from falling asleep in the dark room late at night, it sometimes helped > with the motion sickness scanning for cells. Cell phones are also good for > when your PI is trying to track you down and you're trying to convince him > that even though you aren't in his lab, you actually *are* doing work. > > M > > > On Fri, Jul 12, 2013 at 8:26 AM, Guy Cox <[hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> And why? >> >> -----Original Message----- >> From: Confocal Microscopy List [mailto:[hidden email]] >> On Behalf Of simon walker >> Sent: Friday, 12 July 2013 10:16 PM >> To: [hidden email] >> Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with >> hybrid detectors)... >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Just received on mobile whilst in confocal room.. >> >> Sent from my mobile >> >> On 12 Jul 2013, at 13:12, "Guy Cox" <[hidden email]> wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Can somebody explain why it is so essential to have cellphones in the >> confocal room? >>> >>> >>> Guy >>> >>> -----Original Message----- >>> From: Confocal Microscopy List >>> [mailto:[hidden email]] On Behalf Of >>> [hidden email] >>> Sent: Friday, 12 July 2013 8:27 AM >>> To: [hidden email] >>> Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference >> with hybrid detectors)... >>> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> ***** >>> >>> Heh heh, and it would be useful to control the confocal by mind >> control... >>> >>> On 12/07/13 6:53 AM, "Tim Feinstein" <[hidden email]> wrote: >>> >>>> ***** >>>> To join, leave or search the confocal microscopy listserv, go to: >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>> ***** >>>> >>>> I am a little concerned about the use of aluminum as a blocking agent >>>> here. Not only is aluminum inert as far as magnets are concerned, >>>> but its use to block mind control waves in the radio spectrum is in >>>> fact counterproductive, as shown by a research team at MIT in 2005: >>>> >>>> [...] According to several (apparent) students from MIT who tested >>>> several hat designs, there was "a 30 db amplification at 2.6 Ghz and >>>> a >>>> 20 db amplification at 1.2 Ghz, regardless of the position of the >>>> antenna on the cranium." >>>> >>>> The students note that "the helmets amplify frequency bands that >>>> coincide with those allocated to the US government between 1.2 Ghz and >> 1.4 Ghz. >>>> According to the FCC, These bands are supposedly reserved for 'radio >>>> location' (ie, GPS), and other communications with satellites."[...] >>>> >>>> http://www.theregister.co.uk/2005/11/11/tinfoil_hats_as_government_pl >>>> ot >>>> / >>>> >>>> The original results have since disappeared from the web, >>>> suspiciously enough. Though it may coincide with the student's >>>> personal web page expiring when he graduated I think we can agree >>>> that the episode raises more questions than it answers. >>>> >>>> Cheers, >>>> >>>> TF >>>> >>>> Timothy Feinstein, Ph.D. >>>> >>>> On Jul 11, 2013, at 10:44 AM, Vladimir Zhukarev >>>> <[hidden email]> wrote: >>>> >>>>> ***** >>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>> ***** >>>>> >>>>> Yep, >>>>> No miracles, no aliens here. Electronics is a science of electric >>>>> contacts, as we all know. It takes just one contact with out of >>>>> range stray capacitance, so that with rest of the electric >>>>> transmittance chain this will create a sensitive antenna (receiver) >>>>> for specific radio frequencies. Purely manufacturing issue. Needs >>>>> clean up from the provider. >>>>> Cheers >>>>> Vladimir >>>>> >>>>> Vladimir Zhukarev, Ph.D >>>>> Advanced Microscopy Consulting >>>>> Life Science and Bio-Medical Applications >>>>> >>>>> 'From Image-to Discovery!' >>>>> [hidden email] >>>>> 610.350.1510 www.from-image-to-discovery.com >>>>> >>>>> --------- Original Message --------- Subject: Tin Foil Hats (Was: Re: >>>>> Possible cell phone interference with hybrid detectors)... >>>>> From: "Peter Gabriel Pitrone" <[hidden email]> >>>>> Date: 7/11/13 2:35 am >>>>> To: [hidden email] >>>>> >>>>> ***** >>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>> ***** >>>>> >>>>> Hey All, >>>>> >>>>> I use Aluminum foil to ward off the aliens stealing my thoughts, and >>>>> of course to ward off WiFi and Cellphone signals... So yeah, that >>>>> should help with the HyD detectors. ;-) >>>>> >>>>> Pete >>>>> >>>>> -- >>>>> Peter Gabriel Pitrone - TechRMS >>>>> Microscopy/Imaging Specialist >>>>> Prof. Dr. Pavel Tomancak group >>>>> Max Planck Institute for >>>>> Molecular Biology and Genetics >>>>> Pfotenhauerstr. 108 >>>>> 01307 Dresden >>>>> >>>>> "If a straight line fit is required, obtain only two data points." - >>>>> Anon. >>>>> >>>>> >>>>> On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote: >>>>> <|> ***** >>>>> <|> To join, leave or search the confocal microscopy listserv, go to: >>>>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>> <|> ***** >>>>> <|> >>>>> <|> Hi out there >>>>> <|> >>>>> <|> Same experience for us. We were getting random interference >>>>> lines from <|> our bright field detector on a NikonA1. It was not >>>>> linked to cell phones <|> though but from the cables of another >>>>> system that stands very nearby. >>>>> <|> Shielding the cables with alu foil worked a charm. >>>>> <|> Sylvie >>>>> <|> >>>>> <|> On 9 Jul 2013, at 23:50, "Alexander,Steve" >>>>> <[hidden email]> >>>>> <|> wrote: >>>>> <|> >>>>> <|> ***** >>>>> <|> To join, leave or search the confocal microscopy listserv, go to: >>>>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>> <|> ***** >>>>> <|> >>>>> <|> You could try to shield them with aluminum foil. I tried that to >>>>> find a <|> interference source on a different system and it worked! >>>>> <|> >>>>> <|> Steve Alexander >>>>> <|> >>>>> <|> >>>>> <|> >>>>> <|> -----Original Message----- >>>>> <|> From: Confocal Microscopy List >>>>> [mailto:[hidden email]] >>>>> <|> On Behalf Of Yevgeniy Romin >>>>> <|> Sent: Tuesday, July 09, 2013 4:36 PM <|> To: >>>>> [hidden email] >>>>> <|> Subject: Re: Possible cell phone interference with hybrid >>>>> detectors <|> <|> ***** <|> To join, leave or search the confocal >>>>> microscopy listserv, go to: >>>>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>> <|> ***** >>>>> <|> >>>>> <|> Hi Ben >>>>> <|> >>>>> <|> We have the same problem with one of our sp5 systems that has >>>>> HyDs on it. >>>>> <|> Every once in a while you see a saturated line across the image. >>>>> We have <|> been careful with cell phones around the system, but >>>>> these lines tend to <|> happen anyway every once in a while. We had >>>>> them happen just from a door <|> opening or closing on the room. My >>>>> impression is that while cell phones <|> could be part of the >>>>> problem, they are probably not the only ones. >>>>> There >>>>> <|> could be plenty of equipment or anything else in the room that >>>>> sometimes <|> interferes. We have these lines very rarely, but I >>>>> can't think of a <|> scenario where they are completely guaranteed >>>>> not to show up. >>>>> <|> >>>>> <|> That being said, we have two other inverted systems, an sp5 with >>>>> 3 HyDs <|> and an SP8 with 2 HyDs. We never see any interference >>>>> lines on these <|> systems, no matter what is going on in the room. >>>>> My theory is that the <|> HyDs on the inverted scanheads are hidden >>>>> away towards the rear of the <|> microscope and are perhaps better >>>>> shielded from interference then those <|> on the upright system, >>>>> where they are just sticking out in the open. I <|> communicated >>>>> this theory to Leica, and they said that this could <|> potentially >>>>> be an explanation, but they haven't yet come up with any idea <|> >>>>> for potential shielding of them. >>>>> <|> >>>>> <|> Hope this helps, >>>>> <|> Yevgeniy >>>>> <|> >>>>> <|> >>>>> <|> >>>>> <|> >>>>> <|> Jul 9, 2013, × 5:24 PM, "Alison North" >>>>> <[hidden email]> >>>>> <|> ÎÁÐÉÓÁÌ(Á): >>>>> <|> >>>>> <|>> ***** >>>>> <|>> To join, leave or search the confocal microscopy listserv, go to: >>>>> <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>> <|>> ***** >>>>> <|>> >>>>> <|>> Hi Ben, >>>>> <|>> >>>>> <|>> We were warned by the Leica guys not to use cellphones anywhere >>>>> near <|>> the scan head on the SP8 - in fact it has a great big >>>>> picture of a <|>> cellphone with a line through it on the scanhead! >>>>> - so I guess it is a <|>> well-known problem, not a special "feature" >>>>> of your system. >>>>> <|>> >>>>> <|>> Best wishes, >>>>> <|>> Alison >>>>> <|>> >>>>> <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote: >>>>> <|>>> ***** >>>>> <|>>> To join, leave or search the confocal microscopy listserv, go to: >>>>> <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>> <|>>> ***** >>>>> <|>>> >>>>> <|>>> Hi all, >>>>> <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed >>>>> <|>>> recently that we will get a series of saturated lines during a >>>>> scan <|>>> at random, intermittent moments. Judging by their >>>>> appearance, my <|>>> guess has been that the detectors were picking >>>>> up cellular data <|>>> transmissions. Every time I saw them, >>>>> someone had a cell phone on <|>>> them, and having them move the >>>>> cell phone to the far side of the <|>>> room, or turning the cell >>>>> phone off, stopped the lines from <|>>> occurring. >>>>> <|>>> >>>>> <|>>> I was wondering if anyone else has noticed this with hybrid >>>>> <|>>> detectors, and whether this is normal behavior or something we >>>>> need <|>>> to have looked into. >>>>> <|>>> >>>>> <|>>> Thanks, >>>>> <|>>> Ben Smith >>>>> <|>>> >>>>> <|>>> Benjamin E. Smith, Ph.D. >>>>> <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist >>>>> II <|>>> University of Oklahoma Norman, OK 73019 <|>>> E-mail: >>>>> [hidden email] <|>>> Voice 405-325-4391 <|>>> FAX >>>>> 405-325-7619 <|>>> http://www.microscopy.ou.edu/ <|>> <|>> -- >>>>> <|>> Alison J. North, Ph.D., <|>> Research Associate Professor and >>>>> <|>> Senior Director of the Bio-Imaging Resource Center, The >>>>> Rockefeller <|>> University, <|>> 1230 York Avenue, <|>> New York, >>>>> <|>> NY 10065. >>>>> <|>> Tel: office ++ 212 327 7488 >>>>> <|>> Tel: lab ++ 212 327 7486 >>>>> <|>> Fax: ++ 212 327 7489 >>>>> <|> >>>>> <|> >>>>> <|> >>>>> ==================================================================== >>>>> = >>>>> <|> >>>>> <|> Please note that this e-mail and any files transmitted from <|> >>>>> Memorial Sloan-Kettering Cancer Center may be privileged, <|> >>>>> confidential, <|> and protected from disclosure under applicable law. >>>>> If the reader of <|> this message is not the intended recipient, or >>>>> an employee or agent <|> responsible for delivering this message to >>>>> the intended recipient, <|> you are hereby notified that any >>>>> reading, dissemination, <|> distribution, <|> copying, or other >>>>> use of this communication or any of its attachments <|> is strictly >> prohibited. >>>>> If you have received this communication in <|> error, please notify >>>>> the sender immediately by replying to this <|> message <|> and >>>>> deleting this message, any attachments, and all copies and <|> >>>>> backups <|> from your computer. >>>>> <|> >> The Babraham Institute, Babraham Research Campus, Cambridge CB22 3AT >> Registered Charity No. 1053902. >> The information transmitted in this email is directed only to the >> addressee. If you received this in error, please contact the sender and >> delete this email from your system. The contents of this e-mail are the >> views of the sender and do not necessarily represent the views of the >> Babraham Institute. Full conditions at: www.babraham.ac.uk< >> http://www.babraham.ac.uk/email_disclaimer.html> >> |