Hi all,
    Our Leica SP8 has a pair of HyD detectors, and I've noticed recently that we will get a series of saturated lines during a scan at random, intermittent moments.  Judging by their appearance, my guess has been that the detectors were picking up cellular data transmissions.  Every time I saw them, someone had a cell phone on them, and having them move the cell phone to the far side of the room, or turning the cell phone off, stopped the lines from occurring.  

    I was wondering if anyone else has noticed this with hybrid detectors, and whether this is normal behavior or something we need to have looked into.

Thanks,
    Ben Smith

Benjamin E. Smith, Ph.D.
Samuel Roberts Noble Microscopy Laboratory
Research Scientist II
University of Oklahoma
Norman, OK 73019
E-mail: [hidden email]
Voice   405-325-4391
FAX  405-325-7619
http://www.microscopy.ou.edu/
Alison J. North Alison J. North
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Possible cell phone interference with hybrid detectors

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Ben,

We were warned by the Leica guys not to use cellphones anywhere near the
scan head on the SP8 - in fact it has a great big picture of a cellphone
with a line through it on the scanhead! - so I guess it is a well-known
problem, not a special "feature" of your system.

Best wishes,
Alison

On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi all,
>      Our Leica SP8 has a pair of HyD detectors, and I've noticed recently that we will get a series of saturated lines during a scan at random, intermittent moments.  Judging by their appearance, my guess has been that the detectors were picking up cellular data transmissions.  Every time I saw them, someone had a cell phone on them, and having them move the cell phone to the far side of the room, or turning the cell phone off, stopped the lines from occurring.
>
>      I was wondering if anyone else has noticed this with hybrid detectors, and whether this is normal behavior or something we need to have looked into.
>
> Thanks,
>      Ben Smith
>
> Benjamin E. Smith, Ph.D.
> Samuel Roberts Noble Microscopy Laboratory
> Research Scientist II
> University of Oklahoma
> Norman, OK 73019
> E-mail: [hidden email]
> Voice   405-325-4391
> FAX  405-325-7619
> http://www.microscopy.ou.edu/

--
Alison J. North, Ph.D.,
Research Associate Professor and
Senior Director of the Bio-Imaging Resource Center,
The Rockefeller University,
1230 York Avenue,
New York,
NY 10065.
Tel: office ++ 212 327 7488
Tel: lab     ++ 212 327 7486
Fax:         ++ 212 327 7489
Romin, Yevgeniy/Sloan Kettering Institute Romin, Yevgeniy/Sloan Kettering Institute
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Possible cell phone interference with hybrid detectors

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Ben

We have the same problem with one of our sp5 systems that has HyDs on it. Every once in a while you see a saturated line across the image. We have been careful with cell phones around the system, but these lines tend to happen anyway every once in a while. We had them happen just from a door opening or closing on the room. My impression is that while cell phones could be part of the problem, they are probably not the only ones. There could be plenty of equipment or anything else in the room that sometimes interferes. We have these lines very rarely, but I can't think of a scenario where they are completely guaranteed not to show up.

That being said, we have two other inverted systems, an sp5 with 3 HyDs and an SP8 with 2 HyDs. We never see any interference  lines on these systems, no matter what is going on in the room. My theory is that the HyDs on the inverted scanheads are hidden away towards the rear of the microscope and are perhaps better shielded from interference then those on the upright system, where they are just sticking out in the open. I communicated this theory to Leica, and they said that this could potentially be an explanation, but they haven't yet come up with any idea for potential shielding of them.

Hope this helps,
Yevgeniy




Jul 9, 2013, в 5:24 PM, "Alison North" <[hidden email]> написал(а):

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Ben,
>
> We were warned by the Leica guys not to use cellphones anywhere near the scan head on the SP8 - in fact it has a great big picture of a cellphone with a line through it on the scanhead! - so I guess it is a well-known problem, not a special "feature" of your system.
>
> Best wishes,
> Alison
>
> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote:
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hi all,
>>     Our Leica SP8 has a pair of HyD detectors, and I've noticed recently that we will get a series of saturated lines during a scan at random, intermittent moments.  Judging by their appearance, my guess has been that the detectors were picking up cellular data transmissions.  Every time I saw them, someone had a cell phone on them, and having them move the cell phone to the far side of the room, or turning the cell phone off, stopped the lines from occurring.
>>
>>     I was wondering if anyone else has noticed this with hybrid detectors, and whether this is normal behavior or something we need to have looked into.
>>
>> Thanks,
>>     Ben Smith
>>
>> Benjamin E. Smith, Ph.D.
>> Samuel Roberts Noble Microscopy Laboratory
>> Research Scientist II
>> University of Oklahoma
>> Norman, OK 73019
>> E-mail: [hidden email]
>> Voice   405-325-4391
>> FAX  405-325-7619
>> http://www.microscopy.ou.edu/
>
> --
> Alison J. North, Ph.D.,
> Research Associate Professor and
> Senior Director of the Bio-Imaging Resource Center,
> The Rockefeller University,
> 1230 York Avenue,
> New York,
> NY 10065.
> Tel: office    ++ 212 327 7488
> Tel: lab        ++ 212 327 7486
> Fax:            ++ 212 327 7489
>

 
     =====================================================================
     
     Please note that this e-mail and any files transmitted from
     Memorial Sloan-Kettering Cancer Center may be privileged, confidential,
     and protected from disclosure under applicable law. If the reader of
     this message is not the intended recipient, or an employee or agent
     responsible for delivering this message to the intended recipient,
     you are hereby notified that any reading, dissemination, distribution,
     copying, or other use of this communication or any of its attachments
     is strictly prohibited.  If you have received this communication in
     error, please notify the sender immediately by replying to this message
     and deleting this message, any attachments, and all copies and backups
     from your computer.
Steve Alexander Steve Alexander
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Possible cell phone interference with hybrid detectors

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

You could try to shield them with aluminum foil. I tried that to find a interference source on a different system and it worked!

Steve Alexander



-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Yevgeniy Romin
Sent: Tuesday, July 09, 2013 4:36 PM
To: [hidden email]
Subject: Re: Possible cell phone interference with hybrid detectors

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Ben

We have the same problem with one of our sp5 systems that has HyDs on it. Every once in a while you see a saturated line across the image. We have been careful with cell phones around the system, but these lines tend to happen anyway every once in a while. We had them happen just from a door opening or closing on the room. My impression is that while cell phones could be part of the problem, they are probably not the only ones. There could be plenty of equipment or anything else in the room that sometimes interferes. We have these lines very rarely, but I can't think of a scenario where they are completely guaranteed not to show up.

That being said, we have two other inverted systems, an sp5 with 3 HyDs and an SP8 with 2 HyDs. We never see any interference  lines on these systems, no matter what is going on in the room. My theory is that the HyDs on the inverted scanheads are hidden away towards the rear of the microscope and are perhaps better shielded from interference then those on the upright system, where they are just sticking out in the open. I communicated this theory to Leica, and they said that this could potentially be an explanation, but they haven't yet come up with any idea for potential shielding of them.

Hope this helps,
Yevgeniy




Jul 9, 2013, в 5:24 PM, "Alison North" <[hidden email]> написал(а):

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Ben,
>
> We were warned by the Leica guys not to use cellphones anywhere near the scan head on the SP8 - in fact it has a great big picture of a cellphone with a line through it on the scanhead! - so I guess it is a well-known problem, not a special "feature" of your system.
>
> Best wishes,
> Alison
>
> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote:
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hi all,
>>     Our Leica SP8 has a pair of HyD detectors, and I've noticed recently that we will get a series of saturated lines during a scan at random, intermittent moments.  Judging by their appearance, my guess has been that the detectors were picking up cellular data transmissions.  Every time I saw them, someone had a cell phone on them, and having them move the cell phone to the far side of the room, or turning the cell phone off, stopped the lines from occurring.
>>
>>     I was wondering if anyone else has noticed this with hybrid detectors, and whether this is normal behavior or something we need to have looked into.
>>
>> Thanks,
>>     Ben Smith
>>
>> Benjamin E. Smith, Ph.D.
>> Samuel Roberts Noble Microscopy Laboratory Research Scientist II
>> University of Oklahoma Norman, OK 73019
>> E-mail: [hidden email]
>> Voice   405-325-4391
>> FAX  405-325-7619
>> http://www.microscopy.ou.edu/
>
> --
> Alison J. North, Ph.D.,
> Research Associate Professor and
> Senior Director of the Bio-Imaging Resource Center, The Rockefeller
> University,
> 1230 York Avenue,
> New York,
> NY 10065.
> Tel: office    ++ 212 327 7488
> Tel: lab        ++ 212 327 7486
> Fax:            ++ 212 327 7489
>

 
     =====================================================================
     
     Please note that this e-mail and any files transmitted from
     Memorial Sloan-Kettering Cancer Center may be privileged, confidential,
     and protected from disclosure under applicable law. If the reader of
     this message is not the intended recipient, or an employee or agent
     responsible for delivering this message to the intended recipient,
     you are hereby notified that any reading, dissemination, distribution,
     copying, or other use of this communication or any of its attachments
     is strictly prohibited.  If you have received this communication in
     error, please notify the sender immediately by replying to this message
     and deleting this message, any attachments, and all copies and backups
     from your computer.
Renato Mortara Renato Mortara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Possible cell phone interference with hybrid detectors

In reply to this post by Smith, Benjamin E.
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Ben

We have a similar array of detectors on an SP5: we DO get cell phone (when
they work..) interference and Leica got us some 'no cell' stickers/warnings
to be placed in the room, right when we had the system installed.

Best,

Renato

Renato Mortara
EPM-UNIFESP
www.ecb.epm.br/~ramortara


2013/7/9 Smith, Benjamin E. <[hidden email]>

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi all,
>     Our Leica SP8 has a pair of HyD detectors, and I've noticed recently
> that we will get a series of saturated lines during a scan at random,
> intermittent moments.  Judging by their appearance, my guess has been that
> the detectors were picking up cellular data transmissions.  Every time I
> saw them, someone had a cell phone on them, and having them move the cell
> phone to the far side of the room, or turning the cell phone off, stopped
> the lines from occurring.
>
>     I was wondering if anyone else has noticed this with hybrid detectors,
> and whether this is normal behavior or something we need to have looked
> into.
>
> Thanks,
>     Ben Smith
>
> Benjamin E. Smith, Ph.D.
> Samuel Roberts Noble Microscopy Laboratory
> Research Scientist II
> University of Oklahoma
> Norman, OK 73019
> E-mail: [hidden email]
> Voice   405-325-4391
> FAX  405-325-7619
> http://www.microscopy.ou.edu/
Rosemary.White Rosemary.White
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Possible cell phone interference with hybrid detectors

In reply to this post by Smith, Benjamin E.
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Yes, we were advised this would be a problem, and have a cell phone bucket
outside the door where people have to leave their phones...
cheers,
Rosemary

Dr Rosemary White
CSIRO Plant Industry
GPO Box 1600
Canberra, ACT 2601
Australia

T 61 2 6246 5475
F 61 2 6246 5334
E [hidden email]


On 10/07/13 7:13 AM, "Smith, Benjamin E." <[hidden email]> wrote:

>*****
>To join, leave or search the confocal microscopy listserv, go to:
>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>*****
>
>Hi all,
>    Our Leica SP8 has a pair of HyD detectors, and I've noticed recently
>that we will get a series of saturated lines during a scan at random,
>intermittent moments.  Judging by their appearance, my guess has been
>that the detectors were picking up cellular data transmissions.  Every
>time I saw them, someone had a cell phone on them, and having them move
>the cell phone to the far side of the room, or turning the cell phone
>off, stopped the lines from occurring.
>
>    I was wondering if anyone else has noticed this with hybrid
>detectors, and whether this is normal behavior or something we need to
>have looked into.
>
>Thanks,
>    Ben Smith
>
>Benjamin E. Smith, Ph.D.
>Samuel Roberts Noble Microscopy Laboratory
>Research Scientist II
>University of Oklahoma
>Norman, OK 73019
>E-mail: [hidden email]
>Voice   405-325-4391
>FAX  405-325-7619
>http://www.microscopy.ou.edu/
Sylvie Le Guyader Sylvie Le Guyader
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Possible cell phone interference with hybrid detectors

In reply to this post by Steve Alexander
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi out there

Same experience for us. We were getting random interference lines from our bright field detector on a NikonA1. It was not linked to cell phones though but from the cables of another system that stands very nearby. Shielding the cables with alu foil worked a charm.
Sylvie

On 9 Jul 2013, at 23:50, "Alexander,Steve" <[hidden email]> wrote:

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

You could try to shield them with aluminum foil. I tried that to find a interference source on a different system and it worked!

Steve Alexander



-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Yevgeniy Romin
Sent: Tuesday, July 09, 2013 4:36 PM
To: [hidden email]
Subject: Re: Possible cell phone interference with hybrid detectors

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Ben

We have the same problem with one of our sp5 systems that has HyDs on it. Every once in a while you see a saturated line across the image. We have been careful with cell phones around the system, but these lines tend to happen anyway every once in a while. We had them happen just from a door opening or closing on the room. My impression is that while cell phones could be part of the problem, they are probably not the only ones. There could be plenty of equipment or anything else in the room that sometimes interferes. We have these lines very rarely, but I can't think of a scenario where they are completely guaranteed not to show up.

That being said, we have two other inverted systems, an sp5 with 3 HyDs and an SP8 with 2 HyDs. We never see any interference  lines on these systems, no matter what is going on in the room. My theory is that the HyDs on the inverted scanheads are hidden away towards the rear of the microscope and are perhaps better shielded from interference then those on the upright system, where they are just sticking out in the open. I communicated this theory to Leica, and they said that this could potentially be an explanation, but they haven't yet come up with any idea for potential shielding of them.

Hope this helps,
Yevgeniy




Jul 9, 2013, в 5:24 PM, "Alison North" <[hidden email]> написал(а):

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Ben,
>
> We were warned by the Leica guys not to use cellphones anywhere near the scan head on the SP8 - in fact it has a great big picture of a cellphone with a line through it on the scanhead! - so I guess it is a well-known problem, not a special "feature" of your system.
>
> Best wishes,
> Alison
>
> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote:
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hi all,
>>    Our Leica SP8 has a pair of HyD detectors, and I've noticed recently that we will get a series of saturated lines during a scan at random, intermittent moments.  Judging by their appearance, my guess has been that the detectors were picking up cellular data transmissions.  Every time I saw them, someone had a cell phone on them, and having them move the cell phone to the far side of the room, or turning the cell phone off, stopped the lines from occurring.
>>
>>    I was wondering if anyone else has noticed this with hybrid detectors, and whether this is normal behavior or something we need to have looked into.
>>
>> Thanks,
>>    Ben Smith
>>
>> Benjamin E. Smith, Ph.D.
>> Samuel Roberts Noble Microscopy Laboratory Research Scientist II
>> University of Oklahoma Norman, OK 73019
>> E-mail: [hidden email]
>> Voice   405-325-4391
>> FAX  405-325-7619
>> http://www.microscopy.ou.edu/
>
> --
> Alison J. North, Ph.D.,
> Research Associate Professor and
> Senior Director of the Bio-Imaging Resource Center, The Rockefeller
> University,
> 1230 York Avenue,
> New York,
> NY 10065.
> Tel: office    ++ 212 327 7488
> Tel: lab        ++ 212 327 7486
> Fax:            ++ 212 327 7489


    =====================================================================

    Please note that this e-mail and any files transmitted from
    Memorial Sloan-Kettering Cancer Center may be privileged, confidential,
    and protected from disclosure under applicable law. If the reader of
    this message is not the intended recipient, or an employee or agent
    responsible for delivering this message to the intended recipient,
    you are hereby notified that any reading, dissemination, distribution,
    copying, or other use of this communication or any of its attachments
    is strictly prohibited.  If you have received this communication in
    error, please notify the sender immediately by replying to this message
    and deleting this message, any attachments, and all copies and backups
    from your computer.
mcammer mcammer
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Possible cell phone interference with hybrid detectors

In reply to this post by Renato Mortara
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Lead foil bags, such as the ones we used to use for protecting film from X-rays in airports, wrapped around the PMTs should solve this problem.  Don't know, but maybe even the lighter electrostatic protective bags for EZ Pass, circuit boards, etc. would work.

________________________________________________________
Michael Cammer, Assistant Research Scientist
Microscopy Core, NYU Langone Medical Center & Skirball Institute of Biomolecular Medicine
Cell: (914) 309-3270   Microscopy Lab: (212) 263-7099   Dustin Lab: (212) 263-3208  

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Renato Mortara
Sent: Tuesday, July 09, 2013 5:55 PM
To: [hidden email]
Subject: Re: Possible cell phone interference with hybrid detectors

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Ben

We have a similar array of detectors on an SP5: we DO get cell phone (when they work..) interference and Leica got us some 'no cell' stickers/warnings to be placed in the room, right when we had the system installed.

Best,

Renato

Renato Mortara
EPM-UNIFESP
www.ecb.epm.br/~ramortara


2013/7/9 Smith, Benjamin E. <[hidden email]>

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi all,
>     Our Leica SP8 has a pair of HyD detectors, and I've noticed
> recently that we will get a series of saturated lines during a scan at
> random, intermittent moments.  Judging by their appearance, my guess
> has been that the detectors were picking up cellular data
> transmissions.  Every time I saw them, someone had a cell phone on
> them, and having them move the cell phone to the far side of the room,
> or turning the cell phone off, stopped the lines from occurring.
>
>     I was wondering if anyone else has noticed this with hybrid
> detectors, and whether this is normal behavior or something we need to
> have looked into.
>
> Thanks,
>     Ben Smith
>
> Benjamin E. Smith, Ph.D.
> Samuel Roberts Noble Microscopy Laboratory Research Scientist II
> University of Oklahoma Norman, OK 73019
> E-mail: [hidden email]
> Voice   405-325-4391
> FAX  405-325-7619
> http://www.microscopy.ou.edu/
Peter Gabriel Pitrone Peter Gabriel Pitrone
Reply | Threaded
Open this post in threaded view
♦
♦
|

Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...

In reply to this post by Sylvie Le Guyader
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hey All,

I use Aluminum foil to ward off the aliens stealing my thoughts, and of
course to ward off WiFi and Cellphone signals... So yeah, that should
help with the HyD detectors. ;-)

Pete

--
Peter Gabriel Pitrone - TechRMS
Microscopy/Imaging Specialist
Prof. Dr. Pavel Tomancak group
Max Planck Institute for
Molecular Biology and Genetics
Pfotenhauerstr. 108
01307 Dresden

"If a straight line fit is required, obtain only two data points." - Anon.


On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote:
<|> *****
<|> To join, leave or search the confocal microscopy listserv, go to:
<|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
<|> *****
<|>
<|> Hi out there
<|>
<|> Same experience for us. We were getting random interference lines from
<|> our bright field detector on a NikonA1. It was not linked to cell phones
<|> though but from the cables of another system that stands very nearby.
<|> Shielding the cables with alu foil worked a charm.
<|> Sylvie
<|>
<|> On 9 Jul 2013, at 23:50, "Alexander,Steve" <[hidden email]>
<|> wrote:
<|>
<|> *****
<|> To join, leave or search the confocal microscopy listserv, go to:
<|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
<|> *****
<|>
<|> You could try to shield them with aluminum foil. I tried that to find a
<|> interference source on a different system and it worked!
<|>
<|> Steve Alexander
<|>
<|>
<|>
<|> -----Original Message-----
<|> From: Confocal Microscopy List [mailto:[hidden email]]
<|> On Behalf Of Yevgeniy Romin
<|> Sent: Tuesday, July 09, 2013 4:36 PM
<|> To: [hidden email]
<|> Subject: Re: Possible cell phone interference with hybrid detectors
<|>
<|> *****
<|> To join, leave or search the confocal microscopy listserv, go to:
<|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
<|> *****
<|>
<|> Hi Ben
<|>
<|> We have the same problem with one of our sp5 systems that has HyDs on
it.
<|> Every once in a while you see a saturated line across the image. We have
<|> been careful with cell phones around the system, but these lines tend to
<|> happen anyway every once in a while. We had them happen just from a door
<|> opening or closing on the room. My impression is that while cell phones
<|> could be part of the problem, they are probably not the only ones. There
<|> could be plenty of equipment or anything else in the room that sometimes
<|> interferes. We have these lines very rarely, but I can't think of a
<|> scenario where they are completely guaranteed not to show up.
<|>
<|> That being said, we have two other inverted systems, an sp5 with 3 HyDs
<|> and an SP8 with 2 HyDs. We never see any interference  lines on these
<|> systems, no matter what is going on in the room. My theory is that the
<|> HyDs on the inverted scanheads are hidden away towards the rear of the
<|> microscope and are perhaps better shielded from interference then those
<|> on the upright system, where they are just sticking out in the open. I
<|> communicated this theory to Leica, and they said that this could
<|> potentially be an explanation, but they haven't yet come up with any
idea
<|> for potential shielding of them.
<|>
<|> Hope this helps,
<|> Yevgeniy
<|>
<|>
<|>
<|>
<|> Jul 9, 2013, × 5:24 PM, "Alison North" <[hidden email]>
<|> ÎÁÐÉÓÁÌ(Á):
<|>
<|>> *****
<|>> To join, leave or search the confocal microscopy listserv, go to:
<|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
<|>> *****
<|>>
<|>> Hi Ben,
<|>>
<|>> We were warned by the Leica guys not to use cellphones anywhere near
<|>> the scan head on the SP8 - in fact it has a great big picture of a
<|>> cellphone with a line through it on the scanhead! - so I guess it is a
<|>> well-known problem, not a special "feature" of your system.
<|>>
<|>> Best wishes,
<|>> Alison
<|>>
<|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote:
<|>>> *****
<|>>> To join, leave or search the confocal microscopy listserv, go to:
<|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
<|>>> *****
<|>>>
<|>>> Hi all,
<|>>>    Our Leica SP8 has a pair of HyD detectors, and I've noticed
<|>>> recently that we will get a series of saturated lines during a scan
<|>>> at random, intermittent moments.  Judging by their appearance, my
<|>>> guess has been that the detectors were picking up cellular data
<|>>> transmissions.  Every time I saw them, someone had a cell phone on
<|>>> them, and having them move the cell phone to the far side of the
<|>>> room, or turning the cell phone off, stopped the lines from
<|>>> occurring.
<|>>>
<|>>>    I was wondering if anyone else has noticed this with hybrid
<|>>> detectors, and whether this is normal behavior or something we need
<|>>> to have looked into.
<|>>>
<|>>> Thanks,
<|>>>    Ben Smith
<|>>>
<|>>> Benjamin E. Smith, Ph.D.
<|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist II
<|>>> University of Oklahoma Norman, OK 73019
<|>>> E-mail: [hidden email]
<|>>> Voice   405-325-4391
<|>>> FAX  405-325-7619
<|>>> http://www.microscopy.ou.edu/
<|>>
<|>> --
<|>> Alison J. North, Ph.D.,
<|>> Research Associate Professor and
<|>> Senior Director of the Bio-Imaging Resource Center, The Rockefeller
<|>> University,
<|>> 1230 York Avenue,
<|>> New York,
<|>> NY 10065.
<|>> Tel: office    ++ 212 327 7488
<|>> Tel: lab        ++ 212 327 7486
<|>> Fax:            ++ 212 327 7489
<|>
<|>
<|>    
=====================================================================
<|>
<|>     Please note that this e-mail and any files transmitted from
<|>     Memorial Sloan-Kettering Cancer Center may be privileged,
<|> confidential,
<|>     and protected from disclosure under applicable law. If the reader of
<|>     this message is not the intended recipient, or an employee or agent
<|>     responsible for delivering this message to the intended recipient,
<|>     you are hereby notified that any reading, dissemination,
<|> distribution,
<|>     copying, or other use of this communication or any of its
attachments
<|>     is strictly prohibited.  If you have received this communication in
<|>     error, please notify the sender immediately by replying to this
<|> message
<|>     and deleting this message, any attachments, and all copies and
<|> backups
<|>     from your computer.
<|>
Vladimir Zhukarev Vladimir Zhukarev
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Yep,
No miracles, no aliens here. Electronics is a science of electric contacts, as we all know. It takes just one contact with out of range stray capacitance, so that with rest of the electric transmittance chain this will create a sensitive antenna (receiver) for specific radio frequencies. Purely manufacturing issue. Needs clean up from the provider.
Cheers
Vladimir
 
Vladimir Zhukarev, Ph.D              
Advanced Microscopy Consulting
Life Science and Bio-Medical Applications
 
'From Image-to Discovery!'
[hidden email]  
610.350.1510      www.from-image-to-discovery.com
 
--------- Original Message --------- Subject: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...
From: "Peter Gabriel Pitrone" <[hidden email]>
Date: 7/11/13 2:35 am
To: [hidden email]

*****
 To join, leave or search the confocal microscopy listserv, go to:
 http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
 *****
 
 Hey All,
 
 I use Aluminum foil to ward off the aliens stealing my thoughts, and of
 course to ward off WiFi and Cellphone signals... So yeah, that should
 help with the HyD detectors. ;-)
 
 Pete
 
 --
 Peter Gabriel Pitrone - TechRMS
 Microscopy/Imaging Specialist
 Prof. Dr. Pavel Tomancak group
 Max Planck Institute for
 Molecular Biology and Genetics
 Pfotenhauerstr. 108
 01307 Dresden
 
 "If a straight line fit is required, obtain only two data points." - Anon.
 
 
 On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote:
 <|> *****
 <|> To join, leave or search the confocal microscopy listserv, go to:
 <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
 <|> *****
 <|>
 <|> Hi out there
 <|>
 <|> Same experience for us. We were getting random interference lines from
 <|> our bright field detector on a NikonA1. It was not linked to cell phones
 <|> though but from the cables of another system that stands very nearby.
 <|> Shielding the cables with alu foil worked a charm.
 <|> Sylvie
 <|>
 <|> On 9 Jul 2013, at 23:50, "Alexander,Steve" <[hidden email]>
 <|> wrote:
 <|>
 <|> *****
 <|> To join, leave or search the confocal microscopy listserv, go to:
 <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
 <|> *****
 <|>
 <|> You could try to shield them with aluminum foil. I tried that to find a
 <|> interference source on a different system and it worked!
 <|>
 <|> Steve Alexander
 <|>
 <|>
 <|>
 <|> -----Original Message-----
 <|> From: Confocal Microscopy List [mailto:[hidden email]]
 <|> On Behalf Of Yevgeniy Romin
 <|> Sent: Tuesday, July 09, 2013 4:36 PM
 <|> To: [hidden email]
 <|> Subject: Re: Possible cell phone interference with hybrid detectors
 <|>
 <|> *****
 <|> To join, leave or search the confocal microscopy listserv, go to:
 <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
 <|> *****
 <|>
 <|> Hi Ben
 <|>
 <|> We have the same problem with one of our sp5 systems that has HyDs on
 it.
 <|> Every once in a while you see a saturated line across the image. We have
 <|> been careful with cell phones around the system, but these lines tend to
 <|> happen anyway every once in a while. We had them happen just from a door
 <|> opening or closing on the room. My impression is that while cell phones
 <|> could be part of the problem, they are probably not the only ones. There
 <|> could be plenty of equipment or anything else in the room that sometimes
 <|> interferes. We have these lines very rarely, but I can't think of a
 <|> scenario where they are completely guaranteed not to show up.
 <|>
 <|> That being said, we have two other inverted systems, an sp5 with 3 HyDs
 <|> and an SP8 with 2 HyDs. We never see any interference lines on these
 <|> systems, no matter what is going on in the room. My theory is that the
 <|> HyDs on the inverted scanheads are hidden away towards the rear of the
 <|> microscope and are perhaps better shielded from interference then those
 <|> on the upright system, where they are just sticking out in the open. I
 <|> communicated this theory to Leica, and they said that this could
 <|> potentially be an explanation, but they haven't yet come up with any
 idea
 <|> for potential shielding of them.
 <|>
 <|> Hope this helps,
 <|> Yevgeniy
 <|>
 <|>
 <|>
 <|>
 <|> Jul 9, 2013, &times; 5:24 PM, "Alison North" <[hidden email]>
 <|> &Icirc;&Aacute;&ETH;&Eacute;&Oacute;&Aacute;&Igrave;(&Aacute;):
 <|>
 <|>> *****
 <|>> To join, leave or search the confocal microscopy listserv, go to:
 <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
 <|>> *****
 <|>>
 <|>> Hi Ben,
 <|>>
 <|>> We were warned by the Leica guys not to use cellphones anywhere near
 <|>> the scan head on the SP8 - in fact it has a great big picture of a
 <|>> cellphone with a line through it on the scanhead! - so I guess it is a
 <|>> well-known problem, not a special "feature" of your system.
 <|>>
 <|>> Best wishes,
 <|>> Alison
 <|>>
 <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote:
 <|>>> *****
 <|>>> To join, leave or search the confocal microscopy listserv, go to:
 <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
 <|>>> *****
 <|>>>
 <|>>> Hi all,
 <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed
 <|>>> recently that we will get a series of saturated lines during a scan
 <|>>> at random, intermittent moments. Judging by their appearance, my
 <|>>> guess has been that the detectors were picking up cellular data
 <|>>> transmissions. Every time I saw them, someone had a cell phone on
 <|>>> them, and having them move the cell phone to the far side of the
 <|>>> room, or turning the cell phone off, stopped the lines from
 <|>>> occurring.
 <|>>>
 <|>>> I was wondering if anyone else has noticed this with hybrid
 <|>>> detectors, and whether this is normal behavior or something we need
 <|>>> to have looked into.
 <|>>>
 <|>>> Thanks,
 <|>>> Ben Smith
 <|>>>
 <|>>> Benjamin E. Smith, Ph.D.
 <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist II
 <|>>> University of Oklahoma Norman, OK 73019
 <|>>> E-mail: [hidden email]
 <|>>> Voice 405-325-4391
 <|>>> FAX 405-325-7619
 <|>>> http://www.microscopy.ou.edu/
 <|>>
 <|>> --
 <|>> Alison J. North, Ph.D.,
 <|>> Research Associate Professor and
 <|>> Senior Director of the Bio-Imaging Resource Center, The Rockefeller
 <|>> University,
 <|>> 1230 York Avenue,
 <|>> New York,
 <|>> NY 10065.
 <|>> Tel: office ++ 212 327 7488
 <|>> Tel: lab ++ 212 327 7486
 <|>> Fax: ++ 212 327 7489
 <|>
 <|>
 <|>
 =====================================================================
 <|>
 <|> Please note that this e-mail and any files transmitted from
 <|> Memorial Sloan-Kettering Cancer Center may be privileged,
 <|> confidential,
 <|> and protected from disclosure under applicable law. If the reader of
 <|> this message is not the intended recipient, or an employee or agent
 <|> responsible for delivering this message to the intended recipient,
 <|> you are hereby notified that any reading, dissemination,
 <|> distribution,
 <|> copying, or other use of this communication or any of its
 attachments
 <|> is strictly prohibited. If you have received this communication in
 <|> error, please notify the sender immediately by replying to this
 <|> message
 <|> and deleting this message, any attachments, and all copies and
 <|> backups
 <|> from your computer.
 <|>
Tim Feinstein-2 Tim Feinstein-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I am a little concerned about the use of aluminum as a blocking agent here.  Not only is aluminum inert as far as magnets are concerned, but its use to block mind control waves in the radio spectrum is in fact counterproductive, as shown by a research team at MIT in 2005:

[...] According to several (apparent) students from MIT who tested several hat designs, there was "a 30 db amplification at 2.6 Ghz and a 20 db amplification at 1.2 Ghz, regardless of the position of the antenna on the cranium."

The students note that "the helmets amplify frequency bands that coincide with those allocated to the US government between 1.2 Ghz and 1.4 Ghz. According to the FCC, These bands are supposedly reserved for 'radio location' (ie, GPS), and other communications with satellites."[...]

http://www.theregister.co.uk/2005/11/11/tinfoil_hats_as_government_plot/

The original results have since disappeared from the web, suspiciously enough.  Though it may coincide with the student's personal web
page expiring when he graduated I think we can agree that the episode raises more questions than it answers.  

Cheers,

TF

Timothy Feinstein, Ph.D.

On Jul 11, 2013, at 10:44 AM, Vladimir Zhukarev <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Yep,
> No miracles, no aliens here. Electronics is a science of electric contacts, as we all know. It takes just one contact with out of range stray capacitance, so that with rest of the electric transmittance chain this will create a sensitive antenna (receiver) for specific radio frequencies. Purely manufacturing issue. Needs clean up from the provider.
> Cheers
> Vladimir
>
> Vladimir Zhukarev, Ph.D              
> Advanced Microscopy Consulting
> Life Science and Bio-Medical Applications
>
> 'From Image-to Discovery!'
> [hidden email]  
> 610.350.1510      www.from-image-to-discovery.com
>
> --------- Original Message --------- Subject: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...
> From: "Peter Gabriel Pitrone" <[hidden email]>
> Date: 7/11/13 2:35 am
> To: [hidden email]
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hey All,
>
> I use Aluminum foil to ward off the aliens stealing my thoughts, and of
> course to ward off WiFi and Cellphone signals... So yeah, that should
> help with the HyD detectors. ;-)
>
> Pete
>
> --
> Peter Gabriel Pitrone - TechRMS
> Microscopy/Imaging Specialist
> Prof. Dr. Pavel Tomancak group
> Max Planck Institute for
> Molecular Biology and Genetics
> Pfotenhauerstr. 108
> 01307 Dresden
>
> "If a straight line fit is required, obtain only two data points." - Anon.
>
>
> On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote:
> <|> *****
> <|> To join, leave or search the confocal microscopy listserv, go to:
> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> <|> *****
> <|>
> <|> Hi out there
> <|>
> <|> Same experience for us. We were getting random interference lines from
> <|> our bright field detector on a NikonA1. It was not linked to cell phones
> <|> though but from the cables of another system that stands very nearby.
> <|> Shielding the cables with alu foil worked a charm.
> <|> Sylvie
> <|>
> <|> On 9 Jul 2013, at 23:50, "Alexander,Steve" <[hidden email]>
> <|> wrote:
> <|>
> <|> *****
> <|> To join, leave or search the confocal microscopy listserv, go to:
> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> <|> *****
> <|>
> <|> You could try to shield them with aluminum foil. I tried that to find a
> <|> interference source on a different system and it worked!
> <|>
> <|> Steve Alexander
> <|>
> <|>
> <|>
> <|> -----Original Message-----
> <|> From: Confocal Microscopy List [mailto:[hidden email]]
> <|> On Behalf Of Yevgeniy Romin
> <|> Sent: Tuesday, July 09, 2013 4:36 PM
> <|> To: [hidden email]
> <|> Subject: Re: Possible cell phone interference with hybrid detectors
> <|>
> <|> *****
> <|> To join, leave or search the confocal microscopy listserv, go to:
> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> <|> *****
> <|>
> <|> Hi Ben
> <|>
> <|> We have the same problem with one of our sp5 systems that has HyDs on
> it.
> <|> Every once in a while you see a saturated line across the image. We have
> <|> been careful with cell phones around the system, but these lines tend to
> <|> happen anyway every once in a while. We had them happen just from a door
> <|> opening or closing on the room. My impression is that while cell phones
> <|> could be part of the problem, they are probably not the only ones. There
> <|> could be plenty of equipment or anything else in the room that sometimes
> <|> interferes. We have these lines very rarely, but I can't think of a
> <|> scenario where they are completely guaranteed not to show up.
> <|>
> <|> That being said, we have two other inverted systems, an sp5 with 3 HyDs
> <|> and an SP8 with 2 HyDs. We never see any interference lines on these
> <|> systems, no matter what is going on in the room. My theory is that the
> <|> HyDs on the inverted scanheads are hidden away towards the rear of the
> <|> microscope and are perhaps better shielded from interference then those
> <|> on the upright system, where they are just sticking out in the open. I
> <|> communicated this theory to Leica, and they said that this could
> <|> potentially be an explanation, but they haven't yet come up with any
> idea
> <|> for potential shielding of them.
> <|>
> <|> Hope this helps,
> <|> Yevgeniy
> <|>
> <|>
> <|>
> <|>
> <|> Jul 9, 2013, &times; 5:24 PM, "Alison North" <[hidden email]>
> <|> &Icirc;&Aacute;&ETH;&Eacute;&Oacute;&Aacute;&Igrave;(&Aacute;):
> <|>
> <|>> *****
> <|>> To join, leave or search the confocal microscopy listserv, go to:
> <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> <|>> *****
> <|>>
> <|>> Hi Ben,
> <|>>
> <|>> We were warned by the Leica guys not to use cellphones anywhere near
> <|>> the scan head on the SP8 - in fact it has a great big picture of a
> <|>> cellphone with a line through it on the scanhead! - so I guess it is a
> <|>> well-known problem, not a special "feature" of your system.
> <|>>
> <|>> Best wishes,
> <|>> Alison
> <|>>
> <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote:
> <|>>> *****
> <|>>> To join, leave or search the confocal microscopy listserv, go to:
> <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> <|>>> *****
> <|>>>
> <|>>> Hi all,
> <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed
> <|>>> recently that we will get a series of saturated lines during a scan
> <|>>> at random, intermittent moments. Judging by their appearance, my
> <|>>> guess has been that the detectors were picking up cellular data
> <|>>> transmissions. Every time I saw them, someone had a cell phone on
> <|>>> them, and having them move the cell phone to the far side of the
> <|>>> room, or turning the cell phone off, stopped the lines from
> <|>>> occurring.
> <|>>>
> <|>>> I was wondering if anyone else has noticed this with hybrid
> <|>>> detectors, and whether this is normal behavior or something we need
> <|>>> to have looked into.
> <|>>>
> <|>>> Thanks,
> <|>>> Ben Smith
> <|>>>
> <|>>> Benjamin E. Smith, Ph.D.
> <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist II
> <|>>> University of Oklahoma Norman, OK 73019
> <|>>> E-mail: [hidden email]
> <|>>> Voice 405-325-4391
> <|>>> FAX 405-325-7619
> <|>>> http://www.microscopy.ou.edu/
> <|>>
> <|>> --
> <|>> Alison J. North, Ph.D.,
> <|>> Research Associate Professor and
> <|>> Senior Director of the Bio-Imaging Resource Center, The Rockefeller
> <|>> University,
> <|>> 1230 York Avenue,
> <|>> New York,
> <|>> NY 10065.
> <|>> Tel: office ++ 212 327 7488
> <|>> Tel: lab ++ 212 327 7486
> <|>> Fax: ++ 212 327 7489
> <|>
> <|>
> <|>
> =====================================================================
> <|>
> <|> Please note that this e-mail and any files transmitted from
> <|> Memorial Sloan-Kettering Cancer Center may be privileged,
> <|> confidential,
> <|> and protected from disclosure under applicable law. If the reader of
> <|> this message is not the intended recipient, or an employee or agent
> <|> responsible for delivering this message to the intended recipient,
> <|> you are hereby notified that any reading, dissemination,
> <|> distribution,
> <|> copying, or other use of this communication or any of its
> attachments
> <|> is strictly prohibited. If you have received this communication in
> <|> error, please notify the sender immediately by replying to this
> <|> message
> <|> and deleting this message, any attachments, and all copies and
> <|> backups
> <|> from your computer.
> <|>
Kitty Lee Kitty Lee
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I'm rather concerned with potentially trapping heat and creating noise for the detectors..hope there will be a Leica answer on this.

kitty

> Date: Thu, 11 Jul 2013 16:53:38 -0400
> From: [hidden email]
> Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...
> To: [hidden email]
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> I am a little concerned about the use of aluminum as a blocking agent here.  Not only is aluminum inert as far as magnets are concerned, but its use to block mind control waves in the radio spectrum is in fact counterproductive, as shown by a research team at MIT in 2005:
>
> [...] According to several (apparent) students from MIT who tested several hat designs, there was "a 30 db amplification at 2.6 Ghz and a 20 db amplification at 1.2 Ghz, regardless of the position of the antenna on the cranium."
>
> The students note that "the helmets amplify frequency bands that coincide with those allocated to the US government between 1.2 Ghz and 1.4 Ghz. According to the FCC, These bands are supposedly reserved for 'radio location' (ie, GPS), and other communications with satellites."[...]
>
> http://www.theregister.co.uk/2005/11/11/tinfoil_hats_as_government_plot/
>
> The original results have since disappeared from the web, suspiciously enough.  Though it may coincide with the student's personal web
> page expiring when he graduated I think we can agree that the episode raises more questions than it answers.  
>
> Cheers,
>
> TF
>
> Timothy Feinstein, Ph.D.
>
> On Jul 11, 2013, at 10:44 AM, Vladimir Zhukarev <[hidden email]> wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > *****
> >
> > Yep,
> > No miracles, no aliens here. Electronics is a science of electric contacts, as we all know. It takes just one contact with out of range stray capacitance, so that with rest of the electric transmittance chain this will create a sensitive antenna (receiver) for specific radio frequencies. Purely manufacturing issue. Needs clean up from the provider.
> > Cheers
> > Vladimir
> >
> > Vladimir Zhukarev, Ph.D              
> > Advanced Microscopy Consulting
> > Life Science and Bio-Medical Applications
> >
> > 'From Image-to Discovery!'
> > [hidden email]  
> > 610.350.1510      www.from-image-to-discovery.com
> >
> > --------- Original Message --------- Subject: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...
> > From: "Peter Gabriel Pitrone" <[hidden email]>
> > Date: 7/11/13 2:35 am
> > To: [hidden email]
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > *****
> >
> > Hey All,
> >
> > I use Aluminum foil to ward off the aliens stealing my thoughts, and of
> > course to ward off WiFi and Cellphone signals... So yeah, that should
> > help with the HyD detectors. ;-)
> >
> > Pete
> >
> > --
> > Peter Gabriel Pitrone - TechRMS
> > Microscopy/Imaging Specialist
> > Prof. Dr. Pavel Tomancak group
> > Max Planck Institute for
> > Molecular Biology and Genetics
> > Pfotenhauerstr. 108
> > 01307 Dresden
> >
> > "If a straight line fit is required, obtain only two data points." - Anon.
> >
> >
> > On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote:
> > <|> *****
> > <|> To join, leave or search the confocal microscopy listserv, go to:
> > <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > <|> *****
> > <|>
> > <|> Hi out there
> > <|>
> > <|> Same experience for us. We were getting random interference lines from
> > <|> our bright field detector on a NikonA1. It was not linked to cell phones
> > <|> though but from the cables of another system that stands very nearby.
> > <|> Shielding the cables with alu foil worked a charm.
> > <|> Sylvie
> > <|>
> > <|> On 9 Jul 2013, at 23:50, "Alexander,Steve" <[hidden email]>
> > <|> wrote:
> > <|>
> > <|> *****
> > <|> To join, leave or search the confocal microscopy listserv, go to:
> > <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > <|> *****
> > <|>
> > <|> You could try to shield them with aluminum foil. I tried that to find a
> > <|> interference source on a different system and it worked!
> > <|>
> > <|> Steve Alexander
> > <|>
> > <|>
> > <|>
> > <|> -----Original Message-----
> > <|> From: Confocal Microscopy List [mailto:[hidden email]]
> > <|> On Behalf Of Yevgeniy Romin
> > <|> Sent: Tuesday, July 09, 2013 4:36 PM
> > <|> To: [hidden email]
> > <|> Subject: Re: Possible cell phone interference with hybrid detectors
> > <|>
> > <|> *****
> > <|> To join, leave or search the confocal microscopy listserv, go to:
> > <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > <|> *****
> > <|>
> > <|> Hi Ben
> > <|>
> > <|> We have the same problem with one of our sp5 systems that has HyDs on
> > it.
> > <|> Every once in a while you see a saturated line across the image. We have
> > <|> been careful with cell phones around the system, but these lines tend to
> > <|> happen anyway every once in a while. We had them happen just from a door
> > <|> opening or closing on the room. My impression is that while cell phones
> > <|> could be part of the problem, they are probably not the only ones. There
> > <|> could be plenty of equipment or anything else in the room that sometimes
> > <|> interferes. We have these lines very rarely, but I can't think of a
> > <|> scenario where they are completely guaranteed not to show up.
> > <|>
> > <|> That being said, we have two other inverted systems, an sp5 with 3 HyDs
> > <|> and an SP8 with 2 HyDs. We never see any interference lines on these
> > <|> systems, no matter what is going on in the room. My theory is that the
> > <|> HyDs on the inverted scanheads are hidden away towards the rear of the
> > <|> microscope and are perhaps better shielded from interference then those
> > <|> on the upright system, where they are just sticking out in the open. I
> > <|> communicated this theory to Leica, and they said that this could
> > <|> potentially be an explanation, but they haven't yet come up with any
> > idea
> > <|> for potential shielding of them.
> > <|>
> > <|> Hope this helps,
> > <|> Yevgeniy
> > <|>
> > <|>
> > <|>
> > <|>
> > <|> Jul 9, 2013, &times; 5:24 PM, "Alison North" <[hidden email]>
> > <|> &Icirc;&Aacute;&ETH;&Eacute;&Oacute;&Aacute;&Igrave;(&Aacute;):
> > <|>
> > <|>> *****
> > <|>> To join, leave or search the confocal microscopy listserv, go to:
> > <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > <|>> *****
> > <|>>
> > <|>> Hi Ben,
> > <|>>
> > <|>> We were warned by the Leica guys not to use cellphones anywhere near
> > <|>> the scan head on the SP8 - in fact it has a great big picture of a
> > <|>> cellphone with a line through it on the scanhead! - so I guess it is a
> > <|>> well-known problem, not a special "feature" of your system.
> > <|>>
> > <|>> Best wishes,
> > <|>> Alison
> > <|>>
> > <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote:
> > <|>>> *****
> > <|>>> To join, leave or search the confocal microscopy listserv, go to:
> > <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > <|>>> *****
> > <|>>>
> > <|>>> Hi all,
> > <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed
> > <|>>> recently that we will get a series of saturated lines during a scan
> > <|>>> at random, intermittent moments. Judging by their appearance, my
> > <|>>> guess has been that the detectors were picking up cellular data
> > <|>>> transmissions. Every time I saw them, someone had a cell phone on
> > <|>>> them, and having them move the cell phone to the far side of the
> > <|>>> room, or turning the cell phone off, stopped the lines from
> > <|>>> occurring.
> > <|>>>
> > <|>>> I was wondering if anyone else has noticed this with hybrid
> > <|>>> detectors, and whether this is normal behavior or something we need
> > <|>>> to have looked into.
> > <|>>>
> > <|>>> Thanks,
> > <|>>> Ben Smith
> > <|>>>
> > <|>>> Benjamin E. Smith, Ph.D.
> > <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist II
> > <|>>> University of Oklahoma Norman, OK 73019
> > <|>>> E-mail: [hidden email]
> > <|>>> Voice 405-325-4391
> > <|>>> FAX 405-325-7619
> > <|>>> http://www.microscopy.ou.edu/
> > <|>>
> > <|>> --
> > <|>> Alison J. North, Ph.D.,
> > <|>> Research Associate Professor and
> > <|>> Senior Director of the Bio-Imaging Resource Center, The Rockefeller
> > <|>> University,
> > <|>> 1230 York Avenue,
> > <|>> New York,
> > <|>> NY 10065.
> > <|>> Tel: office ++ 212 327 7488
> > <|>> Tel: lab ++ 212 327 7486
> > <|>> Fax: ++ 212 327 7489
> > <|>
> > <|>
> > <|>
> > =====================================================================
> > <|>
> > <|> Please note that this e-mail and any files transmitted from
> > <|> Memorial Sloan-Kettering Cancer Center may be privileged,
> > <|> confidential,
> > <|> and protected from disclosure under applicable law. If the reader of
> > <|> this message is not the intended recipient, or an employee or agent
> > <|> responsible for delivering this message to the intended recipient,
> > <|> you are hereby notified that any reading, dissemination,
> > <|> distribution,
> > <|> copying, or other use of this communication or any of its
> > attachments
> > <|> is strictly prohibited. If you have received this communication in
> > <|> error, please notify the sender immediately by replying to this
> > <|> message
> > <|> and deleting this message, any attachments, and all copies and
> > <|> backups
> > <|> from your computer.
> > <|>
     
Mike Buchin Mike Buchin
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I think there are fine copper screens/mesh materials that would pass heat
but possibly block rf.  Think Faraday cages.

Mike

Michael Buchin
Stanford Photonics, Inc.
Ph: 650-969-5991


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Kitty Lee
Sent: Thursday, July 11, 2013 2:06 PM
To: [hidden email]
Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with
hybrid detectors)...

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I'm rather concerned with potentially trapping heat and creating noise for
the detectors..hope there will be a Leica answer on this.

kitty

> Date: Thu, 11 Jul 2013 16:53:38 -0400
> From: [hidden email]
> Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with
hybrid detectors)...
> To: [hidden email]
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> I am a little concerned about the use of aluminum as a blocking agent
here.  Not only is aluminum inert as far as magnets are concerned, but its
use to block mind control waves in the radio spectrum is in fact
counterproductive, as shown by a research team at MIT in 2005:
>
> [...] According to several (apparent) students from MIT who tested several
hat designs, there was "a 30 db amplification at 2.6 Ghz and a 20 db
amplification at 1.2 Ghz, regardless of the position of the antenna on the
cranium."

>
> The students note that "the helmets amplify frequency bands that
> coincide with those allocated to the US government between 1.2 Ghz and
> 1.4 Ghz. According to the FCC, These bands are supposedly reserved for
> 'radio location' (ie, GPS), and other communications with
> satellites."[...]
>
> http://www.theregister.co.uk/2005/11/11/tinfoil_hats_as_government_plo
> t/
>
> The original results have since disappeared from the web, suspiciously
> enough.  Though it may coincide with the student's personal web page
expiring when he graduated I think we can agree that the episode raises more
questions than it answers.
>
> Cheers,
>
> TF
>
> Timothy Feinstein, Ph.D.
>
> On Jul 11, 2013, at 10:44 AM, Vladimir Zhukarev
<[hidden email]> wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > *****
> >
> > Yep,
> > No miracles, no aliens here. Electronics is a science of electric
contacts, as we all know. It takes just one contact with out of range stray
capacitance, so that with rest of the electric transmittance chain this will
create a sensitive antenna (receiver) for specific radio frequencies. Purely
manufacturing issue. Needs clean up from the provider.

> > Cheers
> > Vladimir
> >
> > Vladimir Zhukarev, Ph.D              
> > Advanced Microscopy Consulting
> > Life Science and Bio-Medical Applications
> >
> > 'From Image-to Discovery!'
> > [hidden email]  
> > 610.350.1510      www.from-image-to-discovery.com
> >
> > --------- Original Message --------- Subject: Tin Foil Hats (Was: Re:
Possible cell phone interference with hybrid detectors)...

> > From: "Peter Gabriel Pitrone" <[hidden email]>
> > Date: 7/11/13 2:35 am
> > To: [hidden email]
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > *****
> >
> > Hey All,
> >
> > I use Aluminum foil to ward off the aliens stealing my thoughts, and
> > of course to ward off WiFi and Cellphone signals... So yeah, that
> > should help with the HyD detectors. ;-)
> >
> > Pete
> >
> > --
> > Peter Gabriel Pitrone - TechRMS
> > Microscopy/Imaging Specialist
> > Prof. Dr. Pavel Tomancak group
> > Max Planck Institute for
> > Molecular Biology and Genetics
> > Pfotenhauerstr. 108
> > 01307 Dresden
> >
> > "If a straight line fit is required, obtain only two data points." -
Anon.

> >
> >
> > On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote:
> > <|> *****
> > <|> To join, leave or search the confocal microscopy listserv, go to:
> > <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > <|> *****
> > <|>
> > <|> Hi out there
> > <|>
> > <|> Same experience for us. We were getting random interference
> > lines from <|> our bright field detector on a NikonA1. It was not
> > linked to cell phones <|> though but from the cables of another system
that stands very nearby.

> > <|> Shielding the cables with alu foil worked a charm.
> > <|> Sylvie
> > <|>
> > <|> On 9 Jul 2013, at 23:50, "Alexander,Steve"
> > <[hidden email]> <|> wrote:
> > <|>
> > <|> *****
> > <|> To join, leave or search the confocal microscopy listserv, go to:
> > <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > <|> *****
> > <|>
> > <|> You could try to shield them with aluminum foil. I tried that to
> > find a <|> interference source on a different system and it worked!
> > <|>
> > <|> Steve Alexander
> > <|>
> > <|>
> > <|>
> > <|> -----Original Message-----
> > <|> From: Confocal Microscopy List
> > [mailto:[hidden email]]
> > <|> On Behalf Of Yevgeniy Romin
> > <|> Sent: Tuesday, July 09, 2013 4:36 PM <|> To:
> > [hidden email] <|> Subject: Re: Possible cell
> > phone interference with hybrid detectors <|> <|> ***** <|> To join,
> > leave or search the confocal microscopy listserv, go to:
> > <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > <|> *****
> > <|>
> > <|> Hi Ben
> > <|>
> > <|> We have the same problem with one of our sp5 systems that has
> > HyDs on it.
> > <|> Every once in a while you see a saturated line across the image.
> > We have <|> been careful with cell phones around the system, but
> > these lines tend to <|> happen anyway every once in a while. We had
> > them happen just from a door <|> opening or closing on the room. My
> > impression is that while cell phones <|> could be part of the
> > problem, they are probably not the only ones. There <|> could be
> > plenty of equipment or anything else in the room that sometimes <|>
> > interferes. We have these lines very rarely, but I can't think of a <|>
scenario where they are completely guaranteed not to show up.

> > <|>
> > <|> That being said, we have two other inverted systems, an sp5 with
> > 3 HyDs <|> and an SP8 with 2 HyDs. We never see any interference
> > lines on these <|> systems, no matter what is going on in the room.
> > My theory is that the <|> HyDs on the inverted scanheads are hidden
> > away towards the rear of the <|> microscope and are perhaps better
> > shielded from interference then those <|> on the upright system,
> > where they are just sticking out in the open. I <|> communicated
> > this theory to Leica, and they said that this could <|> potentially
> > be an explanation, but they haven't yet come up with any idea <|>
> > for potential shielding of them.
> > <|>
> > <|> Hope this helps,
> > <|> Yevgeniy
> > <|>
> > <|>
> > <|>
> > <|>
> > <|> Jul 9, 2013, &times; 5:24 PM, "Alison North"
> > <[hidden email]> <|>
&Icirc;&Aacute;&ETH;&Eacute;&Oacute;&Aacute;&Igrave;(&Aacute;):

> > <|>
> > <|>> *****
> > <|>> To join, leave or search the confocal microscopy listserv, go to:
> > <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > <|>> *****
> > <|>>
> > <|>> Hi Ben,
> > <|>>
> > <|>> We were warned by the Leica guys not to use cellphones anywhere
> > near <|>> the scan head on the SP8 - in fact it has a great big
> > picture of a <|>> cellphone with a line through it on the scanhead!
> > - so I guess it is a <|>> well-known problem, not a special "feature" of
your system.

> > <|>>
> > <|>> Best wishes,
> > <|>> Alison
> > <|>>
> > <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote:
> > <|>>> *****
> > <|>>> To join, leave or search the confocal microscopy listserv, go to:
> > <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > <|>>> *****
> > <|>>>
> > <|>>> Hi all,
> > <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed
> > <|>>> recently that we will get a series of saturated lines during a
> > scan <|>>> at random, intermittent moments. Judging by their
> > appearance, my <|>>> guess has been that the detectors were picking
> > up cellular data <|>>> transmissions. Every time I saw them, someone
> > had a cell phone on <|>>> them, and having them move the cell phone
> > to the far side of the <|>>> room, or turning the cell phone off,
> > stopped the lines from <|>>> occurring.
> > <|>>>
> > <|>>> I was wondering if anyone else has noticed this with hybrid
> > <|>>> detectors, and whether this is normal behavior or something we
> > need <|>>> to have looked into.
> > <|>>>
> > <|>>> Thanks,
> > <|>>> Ben Smith
> > <|>>>
> > <|>>> Benjamin E. Smith, Ph.D.
> > <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist
> > II <|>>> University of Oklahoma Norman, OK 73019 <|>>> E-mail:
> > [hidden email] <|>>> Voice 405-325-4391 <|>>> FAX
> > 405-325-7619 <|>>> http://www.microscopy.ou.edu/ <|>> <|>> -- <|>>
> > Alison J. North, Ph.D., <|>> Research Associate Professor and <|>>
> > Senior Director of the Bio-Imaging Resource Center, The Rockefeller
> > <|>> University, <|>> 1230 York Avenue, <|>> New York, <|>> NY
> > 10065.
> > <|>> Tel: office ++ 212 327 7488
> > <|>> Tel: lab ++ 212 327 7486
> > <|>> Fax: ++ 212 327 7489
> > <|>
> > <|>
> > <|>
> > ====================================================================
> > =
> > <|>
> > <|> Please note that this e-mail and any files transmitted from <|>
> > Memorial Sloan-Kettering Cancer Center may be privileged, <|>
> > confidential, <|> and protected from disclosure under applicable
> > law. If the reader of <|> this message is not the intended
> > recipient, or an employee or agent <|> responsible for delivering
> > this message to the intended recipient, <|> you are hereby notified
> > that any reading, dissemination, <|> distribution, <|> copying, or
> > other use of this communication or any of its attachments <|> is
> > strictly prohibited. If you have received this communication in <|>
> > error, please notify the sender immediately by replying to this <|>
> > message <|> and deleting this message, any attachments, and all
> > copies and <|> backups <|> from your computer.
> > <|>
     
Rosemary.White Rosemary.White
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...

In reply to this post by Tim Feinstein-2
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Heh heh, and it would be useful to control the confocal by mind control...

On 12/07/13 6:53 AM, "Tim Feinstein" <[hidden email]> wrote:

>*****
>To join, leave or search the confocal microscopy listserv, go to:
>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>*****
>
>I am a little concerned about the use of aluminum as a blocking agent
>here.  Not only is aluminum inert as far as magnets are concerned, but
>its use to block mind control waves in the radio spectrum is in fact
>counterproductive, as shown by a research team at MIT in 2005:
>
>[...] According to several (apparent) students from MIT who tested
>several hat designs, there was "a 30 db amplification at 2.6 Ghz and a 20
>db amplification at 1.2 Ghz, regardless of the position of the antenna on
>the cranium."
>
>The students note that "the helmets amplify frequency bands that coincide
>with those allocated to the US government between 1.2 Ghz and 1.4 Ghz.
>According to the FCC, These bands are supposedly reserved for 'radio
>location' (ie, GPS), and other communications with satellites."[...]
>
>http://www.theregister.co.uk/2005/11/11/tinfoil_hats_as_government_plot/
>
>The original results have since disappeared from the web, suspiciously
>enough.  Though it may coincide with the student's personal web
>page expiring when he graduated I think we can agree that the episode
>raises more questions than it answers.
>
>Cheers,
>
>TF
>
>Timothy Feinstein, Ph.D.
>
>On Jul 11, 2013, at 10:44 AM, Vladimir Zhukarev
><[hidden email]> wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Yep,
>> No miracles, no aliens here. Electronics is a science of electric
>>contacts, as we all know. It takes just one contact with out of range
>>stray capacitance, so that with rest of the electric transmittance chain
>>this will create a sensitive antenna (receiver) for specific radio
>>frequencies. Purely manufacturing issue. Needs clean up from the
>>provider.
>> Cheers
>> Vladimir
>>
>> Vladimir Zhukarev, Ph.D
>> Advanced Microscopy Consulting
>> Life Science and Bio-Medical Applications
>>
>> 'From Image-to Discovery!'
>> [hidden email]
>> 610.350.1510      www.from-image-to-discovery.com
>>
>> --------- Original Message --------- Subject: Tin Foil Hats (Was: Re:
>>Possible cell phone interference with hybrid detectors)...
>> From: "Peter Gabriel Pitrone" <[hidden email]>
>> Date: 7/11/13 2:35 am
>> To: [hidden email]
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hey All,
>>
>> I use Aluminum foil to ward off the aliens stealing my thoughts, and of
>> course to ward off WiFi and Cellphone signals... So yeah, that should
>> help with the HyD detectors. ;-)
>>
>> Pete
>>
>> --
>> Peter Gabriel Pitrone - TechRMS
>> Microscopy/Imaging Specialist
>> Prof. Dr. Pavel Tomancak group
>> Max Planck Institute for
>> Molecular Biology and Genetics
>> Pfotenhauerstr. 108
>> 01307 Dresden
>>
>> "If a straight line fit is required, obtain only two data points." -
>>Anon.
>>
>>
>> On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote:
>> <|> *****
>> <|> To join, leave or search the confocal microscopy listserv, go to:
>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> <|> *****
>> <|>
>> <|> Hi out there
>> <|>
>> <|> Same experience for us. We were getting random interference lines
>>from
>> <|> our bright field detector on a NikonA1. It was not linked to cell
>>phones
>> <|> though but from the cables of another system that stands very
>>nearby.
>> <|> Shielding the cables with alu foil worked a charm.
>> <|> Sylvie
>> <|>
>> <|> On 9 Jul 2013, at 23:50, "Alexander,Steve"
>><[hidden email]>
>> <|> wrote:
>> <|>
>> <|> *****
>> <|> To join, leave or search the confocal microscopy listserv, go to:
>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> <|> *****
>> <|>
>> <|> You could try to shield them with aluminum foil. I tried that to
>>find a
>> <|> interference source on a different system and it worked!
>> <|>
>> <|> Steve Alexander
>> <|>
>> <|>
>> <|>
>> <|> -----Original Message-----
>> <|> From: Confocal Microscopy List
>>[mailto:[hidden email]]
>> <|> On Behalf Of Yevgeniy Romin
>> <|> Sent: Tuesday, July 09, 2013 4:36 PM
>> <|> To: [hidden email]
>> <|> Subject: Re: Possible cell phone interference with hybrid detectors
>> <|>
>> <|> *****
>> <|> To join, leave or search the confocal microscopy listserv, go to:
>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> <|> *****
>> <|>
>> <|> Hi Ben
>> <|>
>> <|> We have the same problem with one of our sp5 systems that has HyDs
>>on
>> it.
>> <|> Every once in a while you see a saturated line across the image. We
>>have
>> <|> been careful with cell phones around the system, but these lines
>>tend to
>> <|> happen anyway every once in a while. We had them happen just from a
>>door
>> <|> opening or closing on the room. My impression is that while cell
>>phones
>> <|> could be part of the problem, they are probably not the only ones.
>>There
>> <|> could be plenty of equipment or anything else in the room that
>>sometimes
>> <|> interferes. We have these lines very rarely, but I can't think of a
>> <|> scenario where they are completely guaranteed not to show up.
>> <|>
>> <|> That being said, we have two other inverted systems, an sp5 with 3
>>HyDs
>> <|> and an SP8 with 2 HyDs. We never see any interference lines on these
>> <|> systems, no matter what is going on in the room. My theory is that
>>the
>> <|> HyDs on the inverted scanheads are hidden away towards the rear of
>>the
>> <|> microscope and are perhaps better shielded from interference then
>>those
>> <|> on the upright system, where they are just sticking out in the
>>open. I
>> <|> communicated this theory to Leica, and they said that this could
>> <|> potentially be an explanation, but they haven't yet come up with any
>> idea
>> <|> for potential shielding of them.
>> <|>
>> <|> Hope this helps,
>> <|> Yevgeniy
>> <|>
>> <|>
>> <|>
>> <|>
>> <|> Jul 9, 2013, &times; 5:24 PM, "Alison North"
>><[hidden email]>
>> <|> &Icirc;&Aacute;&ETH;&Eacute;&Oacute;&Aacute;&Igrave;(&Aacute;):
>> <|>
>> <|>> *****
>> <|>> To join, leave or search the confocal microscopy listserv, go to:
>> <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> <|>> *****
>> <|>>
>> <|>> Hi Ben,
>> <|>>
>> <|>> We were warned by the Leica guys not to use cellphones anywhere
>>near
>> <|>> the scan head on the SP8 - in fact it has a great big picture of a
>> <|>> cellphone with a line through it on the scanhead! - so I guess it
>>is a
>> <|>> well-known problem, not a special "feature" of your system.
>> <|>>
>> <|>> Best wishes,
>> <|>> Alison
>> <|>>
>> <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote:
>> <|>>> *****
>> <|>>> To join, leave or search the confocal microscopy listserv, go to:
>> <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> <|>>> *****
>> <|>>>
>> <|>>> Hi all,
>> <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed
>> <|>>> recently that we will get a series of saturated lines during a
>>scan
>> <|>>> at random, intermittent moments. Judging by their appearance, my
>> <|>>> guess has been that the detectors were picking up cellular data
>> <|>>> transmissions. Every time I saw them, someone had a cell phone on
>> <|>>> them, and having them move the cell phone to the far side of the
>> <|>>> room, or turning the cell phone off, stopped the lines from
>> <|>>> occurring.
>> <|>>>
>> <|>>> I was wondering if anyone else has noticed this with hybrid
>> <|>>> detectors, and whether this is normal behavior or something we
>>need
>> <|>>> to have looked into.
>> <|>>>
>> <|>>> Thanks,
>> <|>>> Ben Smith
>> <|>>>
>> <|>>> Benjamin E. Smith, Ph.D.
>> <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist II
>> <|>>> University of Oklahoma Norman, OK 73019
>> <|>>> E-mail: [hidden email]
>> <|>>> Voice 405-325-4391
>> <|>>> FAX 405-325-7619
>> <|>>> http://www.microscopy.ou.edu/
>> <|>>
>> <|>> --
>> <|>> Alison J. North, Ph.D.,
>> <|>> Research Associate Professor and
>> <|>> Senior Director of the Bio-Imaging Resource Center, The Rockefeller
>> <|>> University,
>> <|>> 1230 York Avenue,
>> <|>> New York,
>> <|>> NY 10065.
>> <|>> Tel: office ++ 212 327 7488
>> <|>> Tel: lab ++ 212 327 7486
>> <|>> Fax: ++ 212 327 7489
>> <|>
>> <|>
>> <|>
>> =====================================================================
>> <|>
>> <|> Please note that this e-mail and any files transmitted from
>> <|> Memorial Sloan-Kettering Cancer Center may be privileged,
>> <|> confidential,
>> <|> and protected from disclosure under applicable law. If the reader of
>> <|> this message is not the intended recipient, or an employee or agent
>> <|> responsible for delivering this message to the intended recipient,
>> <|> you are hereby notified that any reading, dissemination,
>> <|> distribution,
>> <|> copying, or other use of this communication or any of its
>> attachments
>> <|> is strictly prohibited. If you have received this communication in
>> <|> error, please notify the sender immediately by replying to this
>> <|> message
>> <|> and deleting this message, any attachments, and all copies and
>> <|> backups
>> <|> from your computer.
>> <|>
Guy Cox-2 Guy Cox-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Can somebody explain why it is so essential to have cellphones in the confocal room?  

                                                                                                         Guy

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email]
Sent: Friday, 12 July 2013 8:27 AM
To: [hidden email]
Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Heh heh, and it would be useful to control the confocal by mind control...

On 12/07/13 6:53 AM, "Tim Feinstein" <[hidden email]> wrote:

>*****
>To join, leave or search the confocal microscopy listserv, go to:
>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>*****
>
>I am a little concerned about the use of aluminum as a blocking agent
>here.  Not only is aluminum inert as far as magnets are concerned, but
>its use to block mind control waves in the radio spectrum is in fact
>counterproductive, as shown by a research team at MIT in 2005:
>
>[...] According to several (apparent) students from MIT who tested
>several hat designs, there was "a 30 db amplification at 2.6 Ghz and a
>20 db amplification at 1.2 Ghz, regardless of the position of the
>antenna on the cranium."
>
>The students note that "the helmets amplify frequency bands that
>coincide with those allocated to the US government between 1.2 Ghz and 1.4 Ghz.
>According to the FCC, These bands are supposedly reserved for 'radio
>location' (ie, GPS), and other communications with satellites."[...]
>
>http://www.theregister.co.uk/2005/11/11/tinfoil_hats_as_government_plot
>/
>
>The original results have since disappeared from the web, suspiciously
>enough.  Though it may coincide with the student's personal web page
>expiring when he graduated I think we can agree that the episode raises
>more questions than it answers.
>
>Cheers,
>
>TF
>
>Timothy Feinstein, Ph.D.
>
>On Jul 11, 2013, at 10:44 AM, Vladimir Zhukarev
><[hidden email]> wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Yep,
>> No miracles, no aliens here. Electronics is a science of electric
>>contacts, as we all know. It takes just one contact with out of range
>>stray capacitance, so that with rest of the electric transmittance
>>chain this will create a sensitive antenna (receiver) for specific
>>radio frequencies. Purely manufacturing issue. Needs clean up from the
>>provider.
>> Cheers
>> Vladimir
>>
>> Vladimir Zhukarev, Ph.D
>> Advanced Microscopy Consulting
>> Life Science and Bio-Medical Applications
>>
>> 'From Image-to Discovery!'
>> [hidden email]
>> 610.350.1510      www.from-image-to-discovery.com
>>
>> --------- Original Message --------- Subject: Tin Foil Hats (Was: Re:
>>Possible cell phone interference with hybrid detectors)...
>> From: "Peter Gabriel Pitrone" <[hidden email]>
>> Date: 7/11/13 2:35 am
>> To: [hidden email]
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hey All,
>>
>> I use Aluminum foil to ward off the aliens stealing my thoughts, and
>> of course to ward off WiFi and Cellphone signals... So yeah, that
>> should help with the HyD detectors. ;-)
>>
>> Pete
>>
>> --
>> Peter Gabriel Pitrone - TechRMS
>> Microscopy/Imaging Specialist
>> Prof. Dr. Pavel Tomancak group
>> Max Planck Institute for
>> Molecular Biology and Genetics
>> Pfotenhauerstr. 108
>> 01307 Dresden
>>
>> "If a straight line fit is required, obtain only two data points." -
>>Anon.
>>
>>
>> On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote:
>> <|> *****
>> <|> To join, leave or search the confocal microscopy listserv, go to:
>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> <|> *****
>> <|>
>> <|> Hi out there
>> <|>
>> <|> Same experience for us. We were getting random interference lines
>>from  <|> our bright field detector on a NikonA1. It was not linked to
>>cell phones  <|> though but from the cables of another system that
>>stands very nearby.
>> <|> Shielding the cables with alu foil worked a charm.
>> <|> Sylvie
>> <|>
>> <|> On 9 Jul 2013, at 23:50, "Alexander,Steve"
>><[hidden email]>
>> <|> wrote:
>> <|>
>> <|> *****
>> <|> To join, leave or search the confocal microscopy listserv, go to:
>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> <|> *****
>> <|>
>> <|> You could try to shield them with aluminum foil. I tried that to
>>find a  <|> interference source on a different system and it worked!
>> <|>
>> <|> Steve Alexander
>> <|>
>> <|>
>> <|>
>> <|> -----Original Message-----
>> <|> From: Confocal Microscopy List
>>[mailto:[hidden email]]
>> <|> On Behalf Of Yevgeniy Romin
>> <|> Sent: Tuesday, July 09, 2013 4:36 PM  <|> To:
>>[hidden email]
>> <|> Subject: Re: Possible cell phone interference with hybrid
>>detectors  <|>  <|> *****  <|> To join, leave or search the confocal
>>microscopy listserv, go to:
>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> <|> *****
>> <|>
>> <|> Hi Ben
>> <|>
>> <|> We have the same problem with one of our sp5 systems that has
>>HyDs on  it.
>> <|> Every once in a while you see a saturated line across the image.
>>We have  <|> been careful with cell phones around the system, but
>>these lines tend to  <|> happen anyway every once in a while. We had
>>them happen just from a door  <|> opening or closing on the room. My
>>impression is that while cell phones  <|> could be part of the
>>problem, they are probably not the only ones.
>>There
>> <|> could be plenty of equipment or anything else in the room that
>>sometimes  <|> interferes. We have these lines very rarely, but I
>>can't think of a  <|> scenario where they are completely guaranteed
>>not to show up.
>> <|>
>> <|> That being said, we have two other inverted systems, an sp5 with
>>3 HyDs  <|> and an SP8 with 2 HyDs. We never see any interference
>>lines on these  <|> systems, no matter what is going on in the room.
>>My theory is that the  <|> HyDs on the inverted scanheads are hidden
>>away towards the rear of the  <|> microscope and are perhaps better
>>shielded from interference then those  <|> on the upright system,
>>where they are just sticking out in the open. I  <|> communicated this
>>theory to Leica, and they said that this could  <|> potentially be an
>>explanation, but they haven't yet come up with any  idea  <|> for
>>potential shielding of them.
>> <|>
>> <|> Hope this helps,
>> <|> Yevgeniy
>> <|>
>> <|>
>> <|>
>> <|>
>> <|> Jul 9, 2013, &times; 5:24 PM, "Alison North"
>><[hidden email]>
>> <|> &Icirc;&Aacute;&ETH;&Eacute;&Oacute;&Aacute;&Igrave;(&Aacute;):
>> <|>
>> <|>> *****
>> <|>> To join, leave or search the confocal microscopy listserv, go to:
>> <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> <|>> *****
>> <|>>
>> <|>> Hi Ben,
>> <|>>
>> <|>> We were warned by the Leica guys not to use cellphones anywhere
>>near  <|>> the scan head on the SP8 - in fact it has a great big
>>picture of a  <|>> cellphone with a line through it on the scanhead! -
>>so I guess it is a  <|>> well-known problem, not a special "feature"
>>of your system.
>> <|>>
>> <|>> Best wishes,
>> <|>> Alison
>> <|>>
>> <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote:
>> <|>>> *****
>> <|>>> To join, leave or search the confocal microscopy listserv, go to:
>> <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> <|>>> *****
>> <|>>>
>> <|>>> Hi all,
>> <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed  
>><|>>> recently that we will get a series of saturated lines during a
>>scan  <|>>> at random, intermittent moments. Judging by their
>>appearance, my  <|>>> guess has been that the detectors were picking
>>up cellular data  <|>>> transmissions. Every time I saw them, someone
>>had a cell phone on  <|>>> them, and having them move the cell phone
>>to the far side of the  <|>>> room, or turning the cell phone off,
>>stopped the lines from  <|>>> occurring.
>> <|>>>
>> <|>>> I was wondering if anyone else has noticed this with hybrid  
>><|>>> detectors, and whether this is normal behavior or something we
>>need  <|>>> to have looked into.
>> <|>>>
>> <|>>> Thanks,
>> <|>>> Ben Smith
>> <|>>>
>> <|>>> Benjamin E. Smith, Ph.D.
>> <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist
>>II  <|>>> University of Oklahoma Norman, OK 73019  <|>>> E-mail:
>>[hidden email]  <|>>> Voice 405-325-4391  <|>>> FAX
>>405-325-7619  <|>>> http://www.microscopy.ou.edu/  <|>>  <|>> --  <|>>
>>Alison J. North, Ph.D.,  <|>> Research Associate Professor and  <|>>
>>Senior Director of the Bio-Imaging Resource Center, The Rockefeller  
>><|>> University,  <|>> 1230 York Avenue,  <|>> New York,  <|>> NY
>>10065.
>> <|>> Tel: office ++ 212 327 7488
>> <|>> Tel: lab ++ 212 327 7486
>> <|>> Fax: ++ 212 327 7489
>> <|>
>> <|>
>> <|>
>> =====================================================================
>> <|>
>> <|> Please note that this e-mail and any files transmitted from  <|>
>>Memorial Sloan-Kettering Cancer Center may be privileged,  <|>
>>confidential,  <|> and protected from disclosure under applicable law.
>>If the reader of  <|> this message is not the intended recipient, or
>>an employee or agent  <|> responsible for delivering this message to
>>the intended recipient,  <|> you are hereby notified that any reading,
>>dissemination,  <|> distribution,  <|> copying, or other use of this
>>communication or any of its  attachments  <|> is strictly prohibited.
>>If you have received this communication in  <|> error, please notify
>>the sender immediately by replying to this  <|> message  <|> and
>>deleting this message, any attachments, and all copies and  <|>
>>backups  <|> from your computer.
>> <|>
simon walker (BI)-2 simon walker (BI)-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Just received on mobile whilst in confocal room..

Sent from my mobile

On 12 Jul 2013, at 13:12, "Guy Cox" <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Can somebody explain why it is so essential to have cellphones in the confocal room?
>
>                                                                                                         Guy
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of [hidden email]
> Sent: Friday, 12 July 2013 8:27 AM
> To: [hidden email]
> Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Heh heh, and it would be useful to control the confocal by mind control...
>
> On 12/07/13 6:53 AM, "Tim Feinstein" <[hidden email]> wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> I am a little concerned about the use of aluminum as a blocking agent
>> here.  Not only is aluminum inert as far as magnets are concerned, but
>> its use to block mind control waves in the radio spectrum is in fact
>> counterproductive, as shown by a research team at MIT in 2005:
>>
>> [...] According to several (apparent) students from MIT who tested
>> several hat designs, there was "a 30 db amplification at 2.6 Ghz and a
>> 20 db amplification at 1.2 Ghz, regardless of the position of the
>> antenna on the cranium."
>>
>> The students note that "the helmets amplify frequency bands that
>> coincide with those allocated to the US government between 1.2 Ghz and 1.4 Ghz.
>> According to the FCC, These bands are supposedly reserved for 'radio
>> location' (ie, GPS), and other communications with satellites."[...]
>>
>> http://www.theregister.co.uk/2005/11/11/tinfoil_hats_as_government_plot
>> /
>>
>> The original results have since disappeared from the web, suspiciously
>> enough.  Though it may coincide with the student's personal web page
>> expiring when he graduated I think we can agree that the episode raises
>> more questions than it answers.
>>
>> Cheers,
>>
>> TF
>>
>> Timothy Feinstein, Ph.D.
>>
>> On Jul 11, 2013, at 10:44 AM, Vladimir Zhukarev
>> <[hidden email]> wrote:
>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Yep,
>>> No miracles, no aliens here. Electronics is a science of electric
>>> contacts, as we all know. It takes just one contact with out of range
>>> stray capacitance, so that with rest of the electric transmittance
>>> chain this will create a sensitive antenna (receiver) for specific
>>> radio frequencies. Purely manufacturing issue. Needs clean up from the
>>> provider.
>>> Cheers
>>> Vladimir
>>>
>>> Vladimir Zhukarev, Ph.D
>>> Advanced Microscopy Consulting
>>> Life Science and Bio-Medical Applications
>>>
>>> 'From Image-to Discovery!'
>>> [hidden email]
>>> 610.350.1510      www.from-image-to-discovery.com
>>>
>>> --------- Original Message --------- Subject: Tin Foil Hats (Was: Re:
>>> Possible cell phone interference with hybrid detectors)...
>>> From: "Peter Gabriel Pitrone" <[hidden email]>
>>> Date: 7/11/13 2:35 am
>>> To: [hidden email]
>>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Hey All,
>>>
>>> I use Aluminum foil to ward off the aliens stealing my thoughts, and
>>> of course to ward off WiFi and Cellphone signals... So yeah, that
>>> should help with the HyD detectors. ;-)
>>>
>>> Pete
>>>
>>> --
>>> Peter Gabriel Pitrone - TechRMS
>>> Microscopy/Imaging Specialist
>>> Prof. Dr. Pavel Tomancak group
>>> Max Planck Institute for
>>> Molecular Biology and Genetics
>>> Pfotenhauerstr. 108
>>> 01307 Dresden
>>>
>>> "If a straight line fit is required, obtain only two data points." -
>>> Anon.
>>>
>>>
>>> On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote:
>>> <|> *****
>>> <|> To join, leave or search the confocal microscopy listserv, go to:
>>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> <|> *****
>>> <|>
>>> <|> Hi out there
>>> <|>
>>> <|> Same experience for us. We were getting random interference lines
>>> from  <|> our bright field detector on a NikonA1. It was not linked to
>>> cell phones  <|> though but from the cables of another system that
>>> stands very nearby.
>>> <|> Shielding the cables with alu foil worked a charm.
>>> <|> Sylvie
>>> <|>
>>> <|> On 9 Jul 2013, at 23:50, "Alexander,Steve"
>>> <[hidden email]>
>>> <|> wrote:
>>> <|>
>>> <|> *****
>>> <|> To join, leave or search the confocal microscopy listserv, go to:
>>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> <|> *****
>>> <|>
>>> <|> You could try to shield them with aluminum foil. I tried that to
>>> find a  <|> interference source on a different system and it worked!
>>> <|>
>>> <|> Steve Alexander
>>> <|>
>>> <|>
>>> <|>
>>> <|> -----Original Message-----
>>> <|> From: Confocal Microscopy List
>>> [mailto:[hidden email]]
>>> <|> On Behalf Of Yevgeniy Romin
>>> <|> Sent: Tuesday, July 09, 2013 4:36 PM  <|> To:
>>> [hidden email]
>>> <|> Subject: Re: Possible cell phone interference with hybrid
>>> detectors  <|>  <|> *****  <|> To join, leave or search the confocal
>>> microscopy listserv, go to:
>>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> <|> *****
>>> <|>
>>> <|> Hi Ben
>>> <|>
>>> <|> We have the same problem with one of our sp5 systems that has
>>> HyDs on  it.
>>> <|> Every once in a while you see a saturated line across the image.
>>> We have  <|> been careful with cell phones around the system, but
>>> these lines tend to  <|> happen anyway every once in a while. We had
>>> them happen just from a door  <|> opening or closing on the room. My
>>> impression is that while cell phones  <|> could be part of the
>>> problem, they are probably not the only ones.
>>> There
>>> <|> could be plenty of equipment or anything else in the room that
>>> sometimes  <|> interferes. We have these lines very rarely, but I
>>> can't think of a  <|> scenario where they are completely guaranteed
>>> not to show up.
>>> <|>
>>> <|> That being said, we have two other inverted systems, an sp5 with
>>> 3 HyDs  <|> and an SP8 with 2 HyDs. We never see any interference
>>> lines on these  <|> systems, no matter what is going on in the room.
>>> My theory is that the  <|> HyDs on the inverted scanheads are hidden
>>> away towards the rear of the  <|> microscope and are perhaps better
>>> shielded from interference then those  <|> on the upright system,
>>> where they are just sticking out in the open. I  <|> communicated this
>>> theory to Leica, and they said that this could  <|> potentially be an
>>> explanation, but they haven't yet come up with any  idea  <|> for
>>> potential shielding of them.
>>> <|>
>>> <|> Hope this helps,
>>> <|> Yevgeniy
>>> <|>
>>> <|>
>>> <|>
>>> <|>
>>> <|> Jul 9, 2013, &times; 5:24 PM, "Alison North"
>>> <[hidden email]>
>>> <|> &Icirc;&Aacute;&ETH;&Eacute;&Oacute;&Aacute;&Igrave;(&Aacute;):
>>> <|>
>>> <|>> *****
>>> <|>> To join, leave or search the confocal microscopy listserv, go to:
>>> <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> <|>> *****
>>> <|>>
>>> <|>> Hi Ben,
>>> <|>>
>>> <|>> We were warned by the Leica guys not to use cellphones anywhere
>>> near  <|>> the scan head on the SP8 - in fact it has a great big
>>> picture of a  <|>> cellphone with a line through it on the scanhead! -
>>> so I guess it is a  <|>> well-known problem, not a special "feature"
>>> of your system.
>>> <|>>
>>> <|>> Best wishes,
>>> <|>> Alison
>>> <|>>
>>> <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote:
>>> <|>>> *****
>>> <|>>> To join, leave or search the confocal microscopy listserv, go to:
>>> <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> <|>>> *****
>>> <|>>>
>>> <|>>> Hi all,
>>> <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed
>>> <|>>> recently that we will get a series of saturated lines during a
>>> scan  <|>>> at random, intermittent moments. Judging by their
>>> appearance, my  <|>>> guess has been that the detectors were picking
>>> up cellular data  <|>>> transmissions. Every time I saw them, someone
>>> had a cell phone on  <|>>> them, and having them move the cell phone
>>> to the far side of the  <|>>> room, or turning the cell phone off,
>>> stopped the lines from  <|>>> occurring.
>>> <|>>>
>>> <|>>> I was wondering if anyone else has noticed this with hybrid
>>> <|>>> detectors, and whether this is normal behavior or something we
>>> need  <|>>> to have looked into.
>>> <|>>>
>>> <|>>> Thanks,
>>> <|>>> Ben Smith
>>> <|>>>
>>> <|>>> Benjamin E. Smith, Ph.D.
>>> <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist
>>> II  <|>>> University of Oklahoma Norman, OK 73019  <|>>> E-mail:
>>> [hidden email]  <|>>> Voice 405-325-4391  <|>>> FAX
>>> 405-325-7619  <|>>> http://www.microscopy.ou.edu/  <|>>  <|>> --  <|>>
>>> Alison J. North, Ph.D.,  <|>> Research Associate Professor and  <|>>
>>> Senior Director of the Bio-Imaging Resource Center, The Rockefeller
>>> <|>> University,  <|>> 1230 York Avenue,  <|>> New York,  <|>> NY
>>> 10065.
>>> <|>> Tel: office ++ 212 327 7488
>>> <|>> Tel: lab ++ 212 327 7486
>>> <|>> Fax: ++ 212 327 7489
>>> <|>
>>> <|>
>>> <|>
>>> =====================================================================
>>> <|>
>>> <|> Please note that this e-mail and any files transmitted from  <|>
>>> Memorial Sloan-Kettering Cancer Center may be privileged,  <|>
>>> confidential,  <|> and protected from disclosure under applicable law.
>>> If the reader of  <|> this message is not the intended recipient, or
>>> an employee or agent  <|> responsible for delivering this message to
>>> the intended recipient,  <|> you are hereby notified that any reading,
>>> dissemination,  <|> distribution,  <|> copying, or other use of this
>>> communication or any of its  attachments  <|> is strictly prohibited.
>>> If you have received this communication in  <|> error, please notify
>>> the sender immediately by replying to this  <|> message  <|> and
>>> deleting this message, any attachments, and all copies and  <|>
>>> backups  <|> from your computer.
>>> <|>
The Babraham Institute, Babraham Research Campus, Cambridge CB22 3AT Registered Charity No. 1053902.
The information transmitted in this email is directed only to the addressee. If you received this in error, please contact the sender and delete this email from your system. The contents of this e-mail are the views of the sender and do not necessarily represent the views of the Babraham Institute. Full conditions at: www.babraham.ac.uk<http://www.babraham.ac.uk/email_disclaimer.html>
Guy Cox-2 Guy Cox-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

And why?

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of simon walker
Sent: Friday, 12 July 2013 10:16 PM
To: [hidden email]
Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Just received on mobile whilst in confocal room..

Sent from my mobile

On 12 Jul 2013, at 13:12, "Guy Cox" <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Can somebody explain why it is so essential to have cellphones in the confocal room?
>
>                                                                                                        
> Guy
>
> -----Original Message-----
> From: Confocal Microscopy List
> [mailto:[hidden email]] On Behalf Of
> [hidden email]
> Sent: Friday, 12 July 2013 8:27 AM
> To: [hidden email]
> Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Heh heh, and it would be useful to control the confocal by mind control...
>
> On 12/07/13 6:53 AM, "Tim Feinstein" <[hidden email]> wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> I am a little concerned about the use of aluminum as a blocking agent
>> here.  Not only is aluminum inert as far as magnets are concerned,
>> but its use to block mind control waves in the radio spectrum is in
>> fact counterproductive, as shown by a research team at MIT in 2005:
>>
>> [...] According to several (apparent) students from MIT who tested
>> several hat designs, there was "a 30 db amplification at 2.6 Ghz and
>> a
>> 20 db amplification at 1.2 Ghz, regardless of the position of the
>> antenna on the cranium."
>>
>> The students note that "the helmets amplify frequency bands that
>> coincide with those allocated to the US government between 1.2 Ghz and 1.4 Ghz.
>> According to the FCC, These bands are supposedly reserved for 'radio
>> location' (ie, GPS), and other communications with satellites."[...]
>>
>> http://www.theregister.co.uk/2005/11/11/tinfoil_hats_as_government_pl
>> ot
>> /
>>
>> The original results have since disappeared from the web,
>> suspiciously enough.  Though it may coincide with the student's
>> personal web page expiring when he graduated I think we can agree
>> that the episode raises more questions than it answers.
>>
>> Cheers,
>>
>> TF
>>
>> Timothy Feinstein, Ph.D.
>>
>> On Jul 11, 2013, at 10:44 AM, Vladimir Zhukarev
>> <[hidden email]> wrote:
>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Yep,
>>> No miracles, no aliens here. Electronics is a science of electric
>>> contacts, as we all know. It takes just one contact with out of
>>> range stray capacitance, so that with rest of the electric
>>> transmittance chain this will create a sensitive antenna (receiver)
>>> for specific radio frequencies. Purely manufacturing issue. Needs
>>> clean up from the provider.
>>> Cheers
>>> Vladimir
>>>
>>> Vladimir Zhukarev, Ph.D
>>> Advanced Microscopy Consulting
>>> Life Science and Bio-Medical Applications
>>>
>>> 'From Image-to Discovery!'
>>> [hidden email]
>>> 610.350.1510      www.from-image-to-discovery.com
>>>
>>> --------- Original Message --------- Subject: Tin Foil Hats (Was: Re:
>>> Possible cell phone interference with hybrid detectors)...
>>> From: "Peter Gabriel Pitrone" <[hidden email]>
>>> Date: 7/11/13 2:35 am
>>> To: [hidden email]
>>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Hey All,
>>>
>>> I use Aluminum foil to ward off the aliens stealing my thoughts, and
>>> of course to ward off WiFi and Cellphone signals... So yeah, that
>>> should help with the HyD detectors. ;-)
>>>
>>> Pete
>>>
>>> --
>>> Peter Gabriel Pitrone - TechRMS
>>> Microscopy/Imaging Specialist
>>> Prof. Dr. Pavel Tomancak group
>>> Max Planck Institute for
>>> Molecular Biology and Genetics
>>> Pfotenhauerstr. 108
>>> 01307 Dresden
>>>
>>> "If a straight line fit is required, obtain only two data points." -
>>> Anon.
>>>
>>>
>>> On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote:
>>> <|> *****
>>> <|> To join, leave or search the confocal microscopy listserv, go to:
>>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> <|> *****
>>> <|>
>>> <|> Hi out there
>>> <|>
>>> <|> Same experience for us. We were getting random interference
>>> lines from  <|> our bright field detector on a NikonA1. It was not
>>> linked to cell phones  <|> though but from the cables of another
>>> system that stands very nearby.
>>> <|> Shielding the cables with alu foil worked a charm.
>>> <|> Sylvie
>>> <|>
>>> <|> On 9 Jul 2013, at 23:50, "Alexander,Steve"
>>> <[hidden email]>
>>> <|> wrote:
>>> <|>
>>> <|> *****
>>> <|> To join, leave or search the confocal microscopy listserv, go to:
>>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> <|> *****
>>> <|>
>>> <|> You could try to shield them with aluminum foil. I tried that to
>>> find a  <|> interference source on a different system and it worked!
>>> <|>
>>> <|> Steve Alexander
>>> <|>
>>> <|>
>>> <|>
>>> <|> -----Original Message-----
>>> <|> From: Confocal Microscopy List
>>> [mailto:[hidden email]]
>>> <|> On Behalf Of Yevgeniy Romin
>>> <|> Sent: Tuesday, July 09, 2013 4:36 PM  <|> To:
>>> [hidden email]
>>> <|> Subject: Re: Possible cell phone interference with hybrid
>>> detectors  <|>  <|> *****  <|> To join, leave or search the confocal
>>> microscopy listserv, go to:
>>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> <|> *****
>>> <|>
>>> <|> Hi Ben
>>> <|>
>>> <|> We have the same problem with one of our sp5 systems that has
>>> HyDs on  it.
>>> <|> Every once in a while you see a saturated line across the image.
>>> We have  <|> been careful with cell phones around the system, but
>>> these lines tend to  <|> happen anyway every once in a while. We had
>>> them happen just from a door  <|> opening or closing on the room. My
>>> impression is that while cell phones  <|> could be part of the
>>> problem, they are probably not the only ones.
>>> There
>>> <|> could be plenty of equipment or anything else in the room that
>>> sometimes  <|> interferes. We have these lines very rarely, but I
>>> can't think of a  <|> scenario where they are completely guaranteed
>>> not to show up.
>>> <|>
>>> <|> That being said, we have two other inverted systems, an sp5 with
>>> 3 HyDs  <|> and an SP8 with 2 HyDs. We never see any interference
>>> lines on these  <|> systems, no matter what is going on in the room.
>>> My theory is that the  <|> HyDs on the inverted scanheads are hidden
>>> away towards the rear of the  <|> microscope and are perhaps better
>>> shielded from interference then those  <|> on the upright system,
>>> where they are just sticking out in the open. I  <|> communicated
>>> this theory to Leica, and they said that this could  <|> potentially
>>> be an explanation, but they haven't yet come up with any  idea  <|>
>>> for potential shielding of them.
>>> <|>
>>> <|> Hope this helps,
>>> <|> Yevgeniy
>>> <|>
>>> <|>
>>> <|>
>>> <|>
>>> <|> Jul 9, 2013, &times; 5:24 PM, "Alison North"
>>> <[hidden email]>
>>> <|> &Icirc;&Aacute;&ETH;&Eacute;&Oacute;&Aacute;&Igrave;(&Aacute;):
>>> <|>
>>> <|>> *****
>>> <|>> To join, leave or search the confocal microscopy listserv, go to:
>>> <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> <|>> *****
>>> <|>>
>>> <|>> Hi Ben,
>>> <|>>
>>> <|>> We were warned by the Leica guys not to use cellphones anywhere
>>> near  <|>> the scan head on the SP8 - in fact it has a great big
>>> picture of a  <|>> cellphone with a line through it on the scanhead!
>>> - so I guess it is a  <|>> well-known problem, not a special "feature"
>>> of your system.
>>> <|>>
>>> <|>> Best wishes,
>>> <|>> Alison
>>> <|>>
>>> <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote:
>>> <|>>> *****
>>> <|>>> To join, leave or search the confocal microscopy listserv, go to:
>>> <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> <|>>> *****
>>> <|>>>
>>> <|>>> Hi all,
>>> <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed
>>> <|>>> recently that we will get a series of saturated lines during a
>>> scan  <|>>> at random, intermittent moments. Judging by their
>>> appearance, my  <|>>> guess has been that the detectors were picking
>>> up cellular data  <|>>> transmissions. Every time I saw them,
>>> someone had a cell phone on  <|>>> them, and having them move the
>>> cell phone to the far side of the  <|>>> room, or turning the cell
>>> phone off, stopped the lines from  <|>>> occurring.
>>> <|>>>
>>> <|>>> I was wondering if anyone else has noticed this with hybrid
>>> <|>>> detectors, and whether this is normal behavior or something we
>>> need  <|>>> to have looked into.
>>> <|>>>
>>> <|>>> Thanks,
>>> <|>>> Ben Smith
>>> <|>>>
>>> <|>>> Benjamin E. Smith, Ph.D.
>>> <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist
>>> II  <|>>> University of Oklahoma Norman, OK 73019  <|>>> E-mail:
>>> [hidden email]  <|>>> Voice 405-325-4391  <|>>> FAX
>>> 405-325-7619  <|>>> http://www.microscopy.ou.edu/  <|>>  <|>> --  
>>> <|>> Alison J. North, Ph.D.,  <|>> Research Associate Professor and  
>>> <|>> Senior Director of the Bio-Imaging Resource Center, The
>>> Rockefeller <|>> University,  <|>> 1230 York Avenue,  <|>> New York,  
>>> <|>> NY 10065.
>>> <|>> Tel: office ++ 212 327 7488
>>> <|>> Tel: lab ++ 212 327 7486
>>> <|>> Fax: ++ 212 327 7489
>>> <|>
>>> <|>
>>> <|>
>>> ====================================================================
>>> =
>>> <|>
>>> <|> Please note that this e-mail and any files transmitted from  <|>
>>> Memorial Sloan-Kettering Cancer Center may be privileged,  <|>
>>> confidential,  <|> and protected from disclosure under applicable law.
>>> If the reader of  <|> this message is not the intended recipient, or
>>> an employee or agent  <|> responsible for delivering this message to
>>> the intended recipient,  <|> you are hereby notified that any
>>> reading, dissemination,  <|> distribution,  <|> copying, or other
>>> use of this communication or any of its  attachments  <|> is strictly prohibited.
>>> If you have received this communication in  <|> error, please notify
>>> the sender immediately by replying to this  <|> message  <|> and
>>> deleting this message, any attachments, and all copies and  <|>
>>> backups  <|> from your computer.
>>> <|>
The Babraham Institute, Babraham Research Campus, Cambridge CB22 3AT Registered Charity No. 1053902.
The information transmitted in this email is directed only to the addressee. If you received this in error, please contact the sender and delete this email from your system. The contents of this e-mail are the views of the sender and do not necessarily represent the views of the Babraham Institute. Full conditions at: www.babraham.ac.uk<http://www.babraham.ac.uk/email_disclaimer.html>
MJ Tomaszewski MJ Tomaszewski
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I once got locked in a confocal room because the inside door knob fell off.
The sad thing is that the room didn't have a working phone yet, nor did it
get cell reception. I had to wait for someone to walk by, get their
attention, then get them to find someone that had card key access to the
room to let me out. This happened twice. Then I started bringing a tool to
the room to reattach the door knob.

I would imagine that most people have cell phones that play music now. I
would often listen to music while on the scopes, not only did it keep me
from falling asleep in the dark room late at night, it sometimes helped
with the motion sickness scanning for cells. Cell phones are also good for
when your PI is trying to track you down and you're trying to convince him
that even though you aren't in his lab, you actually *are* doing work.

M


On Fri, Jul 12, 2013 at 8:26 AM, Guy Cox <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> And why?
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]]
> On Behalf Of simon walker
> Sent: Friday, 12 July 2013 10:16 PM
> To: [hidden email]
> Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with
> hybrid detectors)...
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Just received on mobile whilst in confocal room..
>
> Sent from my mobile
>
> On 12 Jul 2013, at 13:12, "Guy Cox" <[hidden email]> wrote:
>
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > *****
> >
> > Can somebody explain why it is so essential to have cellphones in the
> confocal room?
> >
> >
> > Guy
> >
> > -----Original Message-----
> > From: Confocal Microscopy List
> > [mailto:[hidden email]] On Behalf Of
> > [hidden email]
> > Sent: Friday, 12 July 2013 8:27 AM
> > To: [hidden email]
> > Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference
> with hybrid detectors)...
> >
> > *****
> > To join, leave or search the confocal microscopy listserv, go to:
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> > *****
> >
> > Heh heh, and it would be useful to control the confocal by mind
> control...
> >
> > On 12/07/13 6:53 AM, "Tim Feinstein" <[hidden email]> wrote:
> >
> >> *****
> >> To join, leave or search the confocal microscopy listserv, go to:
> >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >> *****
> >>
> >> I am a little concerned about the use of aluminum as a blocking agent
> >> here.  Not only is aluminum inert as far as magnets are concerned,
> >> but its use to block mind control waves in the radio spectrum is in
> >> fact counterproductive, as shown by a research team at MIT in 2005:
> >>
> >> [...] According to several (apparent) students from MIT who tested
> >> several hat designs, there was "a 30 db amplification at 2.6 Ghz and
> >> a
> >> 20 db amplification at 1.2 Ghz, regardless of the position of the
> >> antenna on the cranium."
> >>
> >> The students note that "the helmets amplify frequency bands that
> >> coincide with those allocated to the US government between 1.2 Ghz and
> 1.4 Ghz.
> >> According to the FCC, These bands are supposedly reserved for 'radio
> >> location' (ie, GPS), and other communications with satellites."[...]
> >>
> >> http://www.theregister.co.uk/2005/11/11/tinfoil_hats_as_government_pl
> >> ot
> >> /
> >>
> >> The original results have since disappeared from the web,
> >> suspiciously enough.  Though it may coincide with the student's
> >> personal web page expiring when he graduated I think we can agree
> >> that the episode raises more questions than it answers.
> >>
> >> Cheers,
> >>
> >> TF
> >>
> >> Timothy Feinstein, Ph.D.
> >>
> >> On Jul 11, 2013, at 10:44 AM, Vladimir Zhukarev
> >> <[hidden email]> wrote:
> >>
> >>> *****
> >>> To join, leave or search the confocal microscopy listserv, go to:
> >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >>> *****
> >>>
> >>> Yep,
> >>> No miracles, no aliens here. Electronics is a science of electric
> >>> contacts, as we all know. It takes just one contact with out of
> >>> range stray capacitance, so that with rest of the electric
> >>> transmittance chain this will create a sensitive antenna (receiver)
> >>> for specific radio frequencies. Purely manufacturing issue. Needs
> >>> clean up from the provider.
> >>> Cheers
> >>> Vladimir
> >>>
> >>> Vladimir Zhukarev, Ph.D
> >>> Advanced Microscopy Consulting
> >>> Life Science and Bio-Medical Applications
> >>>
> >>> 'From Image-to Discovery!'
> >>> [hidden email]
> >>> 610.350.1510      www.from-image-to-discovery.com
> >>>
> >>> --------- Original Message --------- Subject: Tin Foil Hats (Was: Re:
> >>> Possible cell phone interference with hybrid detectors)...
> >>> From: "Peter Gabriel Pitrone" <[hidden email]>
> >>> Date: 7/11/13 2:35 am
> >>> To: [hidden email]
> >>>
> >>> *****
> >>> To join, leave or search the confocal microscopy listserv, go to:
> >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >>> *****
> >>>
> >>> Hey All,
> >>>
> >>> I use Aluminum foil to ward off the aliens stealing my thoughts, and
> >>> of course to ward off WiFi and Cellphone signals... So yeah, that
> >>> should help with the HyD detectors. ;-)
> >>>
> >>> Pete
> >>>
> >>> --
> >>> Peter Gabriel Pitrone - TechRMS
> >>> Microscopy/Imaging Specialist
> >>> Prof. Dr. Pavel Tomancak group
> >>> Max Planck Institute for
> >>> Molecular Biology and Genetics
> >>> Pfotenhauerstr. 108
> >>> 01307 Dresden
> >>>
> >>> "If a straight line fit is required, obtain only two data points." -
> >>> Anon.
> >>>
> >>>
> >>> On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote:
> >>> <|> *****
> >>> <|> To join, leave or search the confocal microscopy listserv, go to:
> >>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >>> <|> *****
> >>> <|>
> >>> <|> Hi out there
> >>> <|>
> >>> <|> Same experience for us. We were getting random interference
> >>> lines from  <|> our bright field detector on a NikonA1. It was not
> >>> linked to cell phones  <|> though but from the cables of another
> >>> system that stands very nearby.
> >>> <|> Shielding the cables with alu foil worked a charm.
> >>> <|> Sylvie
> >>> <|>
> >>> <|> On 9 Jul 2013, at 23:50, "Alexander,Steve"
> >>> <[hidden email]>
> >>> <|> wrote:
> >>> <|>
> >>> <|> *****
> >>> <|> To join, leave or search the confocal microscopy listserv, go to:
> >>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >>> <|> *****
> >>> <|>
> >>> <|> You could try to shield them with aluminum foil. I tried that to
> >>> find a  <|> interference source on a different system and it worked!
> >>> <|>
> >>> <|> Steve Alexander
> >>> <|>
> >>> <|>
> >>> <|>
> >>> <|> -----Original Message-----
> >>> <|> From: Confocal Microscopy List
> >>> [mailto:[hidden email]]
> >>> <|> On Behalf Of Yevgeniy Romin
> >>> <|> Sent: Tuesday, July 09, 2013 4:36 PM  <|> To:
> >>> [hidden email]
> >>> <|> Subject: Re: Possible cell phone interference with hybrid
> >>> detectors  <|>  <|> *****  <|> To join, leave or search the confocal
> >>> microscopy listserv, go to:
> >>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >>> <|> *****
> >>> <|>
> >>> <|> Hi Ben
> >>> <|>
> >>> <|> We have the same problem with one of our sp5 systems that has
> >>> HyDs on  it.
> >>> <|> Every once in a while you see a saturated line across the image.
> >>> We have  <|> been careful with cell phones around the system, but
> >>> these lines tend to  <|> happen anyway every once in a while. We had
> >>> them happen just from a door  <|> opening or closing on the room. My
> >>> impression is that while cell phones  <|> could be part of the
> >>> problem, they are probably not the only ones.
> >>> There
> >>> <|> could be plenty of equipment or anything else in the room that
> >>> sometimes  <|> interferes. We have these lines very rarely, but I
> >>> can't think of a  <|> scenario where they are completely guaranteed
> >>> not to show up.
> >>> <|>
> >>> <|> That being said, we have two other inverted systems, an sp5 with
> >>> 3 HyDs  <|> and an SP8 with 2 HyDs. We never see any interference
> >>> lines on these  <|> systems, no matter what is going on in the room.
> >>> My theory is that the  <|> HyDs on the inverted scanheads are hidden
> >>> away towards the rear of the  <|> microscope and are perhaps better
> >>> shielded from interference then those  <|> on the upright system,
> >>> where they are just sticking out in the open. I  <|> communicated
> >>> this theory to Leica, and they said that this could  <|> potentially
> >>> be an explanation, but they haven't yet come up with any  idea  <|>
> >>> for potential shielding of them.
> >>> <|>
> >>> <|> Hope this helps,
> >>> <|> Yevgeniy
> >>> <|>
> >>> <|>
> >>> <|>
> >>> <|>
> >>> <|> Jul 9, 2013, &times; 5:24 PM, "Alison North"
> >>> <[hidden email]>
> >>> <|> &Icirc;&Aacute;&ETH;&Eacute;&Oacute;&Aacute;&Igrave;(&Aacute;):
> >>> <|>
> >>> <|>> *****
> >>> <|>> To join, leave or search the confocal microscopy listserv, go to:
> >>> <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >>> <|>> *****
> >>> <|>>
> >>> <|>> Hi Ben,
> >>> <|>>
> >>> <|>> We were warned by the Leica guys not to use cellphones anywhere
> >>> near  <|>> the scan head on the SP8 - in fact it has a great big
> >>> picture of a  <|>> cellphone with a line through it on the scanhead!
> >>> - so I guess it is a  <|>> well-known problem, not a special "feature"
> >>> of your system.
> >>> <|>>
> >>> <|>> Best wishes,
> >>> <|>> Alison
> >>> <|>>
> >>> <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote:
> >>> <|>>> *****
> >>> <|>>> To join, leave or search the confocal microscopy listserv, go to:
> >>> <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> >>> <|>>> *****
> >>> <|>>>
> >>> <|>>> Hi all,
> >>> <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed
> >>> <|>>> recently that we will get a series of saturated lines during a
> >>> scan  <|>>> at random, intermittent moments. Judging by their
> >>> appearance, my  <|>>> guess has been that the detectors were picking
> >>> up cellular data  <|>>> transmissions. Every time I saw them,
> >>> someone had a cell phone on  <|>>> them, and having them move the
> >>> cell phone to the far side of the  <|>>> room, or turning the cell
> >>> phone off, stopped the lines from  <|>>> occurring.
> >>> <|>>>
> >>> <|>>> I was wondering if anyone else has noticed this with hybrid
> >>> <|>>> detectors, and whether this is normal behavior or something we
> >>> need  <|>>> to have looked into.
> >>> <|>>>
> >>> <|>>> Thanks,
> >>> <|>>> Ben Smith
> >>> <|>>>
> >>> <|>>> Benjamin E. Smith, Ph.D.
> >>> <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist
> >>> II  <|>>> University of Oklahoma Norman, OK 73019  <|>>> E-mail:
> >>> [hidden email]  <|>>> Voice 405-325-4391  <|>>> FAX
> >>> 405-325-7619  <|>>> http://www.microscopy.ou.edu/  <|>>  <|>> --
> >>> <|>> Alison J. North, Ph.D.,  <|>> Research Associate Professor and
> >>> <|>> Senior Director of the Bio-Imaging Resource Center, The
> >>> Rockefeller <|>> University,  <|>> 1230 York Avenue,  <|>> New York,
> >>> <|>> NY 10065.
> >>> <|>> Tel: office ++ 212 327 7488
> >>> <|>> Tel: lab ++ 212 327 7486
> >>> <|>> Fax: ++ 212 327 7489
> >>> <|>
> >>> <|>
> >>> <|>
> >>> ====================================================================
> >>> =
> >>> <|>
> >>> <|> Please note that this e-mail and any files transmitted from  <|>
> >>> Memorial Sloan-Kettering Cancer Center may be privileged,  <|>
> >>> confidential,  <|> and protected from disclosure under applicable law.
> >>> If the reader of  <|> this message is not the intended recipient, or
> >>> an employee or agent  <|> responsible for delivering this message to
> >>> the intended recipient,  <|> you are hereby notified that any
> >>> reading, dissemination,  <|> distribution,  <|> copying, or other
> >>> use of this communication or any of its  attachments  <|> is strictly
> prohibited.
> >>> If you have received this communication in  <|> error, please notify
> >>> the sender immediately by replying to this  <|> message  <|> and
> >>> deleting this message, any attachments, and all copies and  <|>
> >>> backups  <|> from your computer.
> >>> <|>
> The Babraham Institute, Babraham Research Campus, Cambridge CB22 3AT
> Registered Charity No. 1053902.
> The information transmitted in this email is directed only to the
> addressee. If you received this in error, please contact the sender and
> delete this email from your system. The contents of this e-mail are the
> views of the sender and do not necessarily represent the views of the
> Babraham Institute. Full conditions at: www.babraham.ac.uk<
> http://www.babraham.ac.uk/email_disclaimer.html>
>
fujisaws fujisaws
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I have a related question.

How is WiFi different from cell phones?  In our institution, there is strong WiFi signal everywhere, including our confocal rooms. Would that cause similar interference??

Also, we have MDs on call using the confocals. They are obliged to carry cell phones all the time. We do ask them to not to make calls in the room, but these things are constantly receiving and sending data.


Sho Fujisawa
Molecular Cytology Core Facility
Memorial Sloan-Kettering Cancer Center
415 East 68th St. ZRC-1840
New York, NY 10065
(646) 888-2186, x2186

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of MJ Tomaszewski
Sent: Friday, July 12, 2013 8:41 AM
To: [hidden email]
Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I once got locked in a confocal room because the inside door knob fell off.
The sad thing is that the room didn't have a working phone yet, nor did it get cell reception. I had to wait for someone to walk by, get their attention, then get them to find someone that had card key access to the room to let me out. This happened twice. Then I started bringing a tool to the room to reattach the door knob.

I would imagine that most people have cell phones that play music now. I would often listen to music while on the scopes, not only did it keep me from falling asleep in the dark room late at night, it sometimes helped with the motion sickness scanning for cells. Cell phones are also good for when your PI is trying to track you down and you're trying to convince him that even though you aren't in his lab, you actually *are* doing work.

M




 
     =====================================================================
     
     Please note that this e-mail and any files transmitted from
     Memorial Sloan-Kettering Cancer Center may be privileged, confidential,
     and protected from disclosure under applicable law. If the reader of
     this message is not the intended recipient, or an employee or agent
     responsible for delivering this message to the intended recipient,
     you are hereby notified that any reading, dissemination, distribution,
     copying, or other use of this communication or any of its attachments
     is strictly prohibited.  If you have received this communication in
     error, please notify the sender immediately by replying to this message
     and deleting this message, any attachments, and all copies and backups
     from your computer.
Chris Tully Chris Tully
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Tin Foil Hats (Was: Re: Possible cell phone interference with hybrid detectors)...

In reply to this post by MJ Tomaszewski
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I will strongly second the idea of a Faraday cage is cell phone
signals are causing that much interference. I would even push back
against the manufacturer to demand that they retro fit the scope with
the necessary shielding. They probably won't but it never hurts to
try.

I say this because I remember back in the 90's seeing sign all over
hospitals to turn off personal cell phones because they interfered
with patient care equipment. More recently though those signs have
disappeared as equipment has been upgraded or replaced.

Back to the Faraday cage idea though: if it is effective enough to
shield the experiments looking at electrical signals of mouse muscles
then it should shield a scope from most external interference. Just
make sure that it fully encloses the components that are picking up
the interfering signals. And ground it!

Chris Tully, M.S., Image Analysis Expert
t 240.475.9753 f 419.831.0527 | [hidden email]

Sent from my iPhone please excuse typos.

On Jul 12, 2013, at 8:42 AM, MJ Tomaszewski <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> I once got locked in a confocal room because the inside door knob fell off.
> The sad thing is that the room didn't have a working phone yet, nor did it
> get cell reception. I had to wait for someone to walk by, get their
> attention, then get them to find someone that had card key access to the
> room to let me out. This happened twice. Then I started bringing a tool to
> the room to reattach the door knob.
>
> I would imagine that most people have cell phones that play music now. I
> would often listen to music while on the scopes, not only did it keep me
> from falling asleep in the dark room late at night, it sometimes helped
> with the motion sickness scanning for cells. Cell phones are also good for
> when your PI is trying to track you down and you're trying to convince him
> that even though you aren't in his lab, you actually *are* doing work.
>
> M
>
>
> On Fri, Jul 12, 2013 at 8:26 AM, Guy Cox <[hidden email]> wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> And why?
>>
>> -----Original Message-----
>> From: Confocal Microscopy List [mailto:[hidden email]]
>> On Behalf Of simon walker
>> Sent: Friday, 12 July 2013 10:16 PM
>> To: [hidden email]
>> Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference with
>> hybrid detectors)...
>>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Just received on mobile whilst in confocal room..
>>
>> Sent from my mobile
>>
>> On 12 Jul 2013, at 13:12, "Guy Cox" <[hidden email]> wrote:
>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Can somebody explain why it is so essential to have cellphones in the
>> confocal room?
>>>
>>>
>>> Guy
>>>
>>> -----Original Message-----
>>> From: Confocal Microscopy List
>>> [mailto:[hidden email]] On Behalf Of
>>> [hidden email]
>>> Sent: Friday, 12 July 2013 8:27 AM
>>> To: [hidden email]
>>> Subject: Re: Tin Foil Hats (Was: Re: Possible cell phone interference
>> with hybrid detectors)...
>>>
>>> *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> Heh heh, and it would be useful to control the confocal by mind
>> control...
>>>
>>> On 12/07/13 6:53 AM, "Tim Feinstein" <[hidden email]> wrote:
>>>
>>>> *****
>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>> *****
>>>>
>>>> I am a little concerned about the use of aluminum as a blocking agent
>>>> here.  Not only is aluminum inert as far as magnets are concerned,
>>>> but its use to block mind control waves in the radio spectrum is in
>>>> fact counterproductive, as shown by a research team at MIT in 2005:
>>>>
>>>> [...] According to several (apparent) students from MIT who tested
>>>> several hat designs, there was "a 30 db amplification at 2.6 Ghz and
>>>> a
>>>> 20 db amplification at 1.2 Ghz, regardless of the position of the
>>>> antenna on the cranium."
>>>>
>>>> The students note that "the helmets amplify frequency bands that
>>>> coincide with those allocated to the US government between 1.2 Ghz and
>> 1.4 Ghz.
>>>> According to the FCC, These bands are supposedly reserved for 'radio
>>>> location' (ie, GPS), and other communications with satellites."[...]
>>>>
>>>> http://www.theregister.co.uk/2005/11/11/tinfoil_hats_as_government_pl
>>>> ot
>>>> /
>>>>
>>>> The original results have since disappeared from the web,
>>>> suspiciously enough.  Though it may coincide with the student's
>>>> personal web page expiring when he graduated I think we can agree
>>>> that the episode raises more questions than it answers.
>>>>
>>>> Cheers,
>>>>
>>>> TF
>>>>
>>>> Timothy Feinstein, Ph.D.
>>>>
>>>> On Jul 11, 2013, at 10:44 AM, Vladimir Zhukarev
>>>> <[hidden email]> wrote:
>>>>
>>>>> *****
>>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>>> *****
>>>>>
>>>>> Yep,
>>>>> No miracles, no aliens here. Electronics is a science of electric
>>>>> contacts, as we all know. It takes just one contact with out of
>>>>> range stray capacitance, so that with rest of the electric
>>>>> transmittance chain this will create a sensitive antenna (receiver)
>>>>> for specific radio frequencies. Purely manufacturing issue. Needs
>>>>> clean up from the provider.
>>>>> Cheers
>>>>> Vladimir
>>>>>
>>>>> Vladimir Zhukarev, Ph.D
>>>>> Advanced Microscopy Consulting
>>>>> Life Science and Bio-Medical Applications
>>>>>
>>>>> 'From Image-to Discovery!'
>>>>> [hidden email]
>>>>> 610.350.1510      www.from-image-to-discovery.com
>>>>>
>>>>> --------- Original Message --------- Subject: Tin Foil Hats (Was: Re:
>>>>> Possible cell phone interference with hybrid detectors)...
>>>>> From: "Peter Gabriel Pitrone" <[hidden email]>
>>>>> Date: 7/11/13 2:35 am
>>>>> To: [hidden email]
>>>>>
>>>>> *****
>>>>> To join, leave or search the confocal microscopy listserv, go to:
>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>>> *****
>>>>>
>>>>> Hey All,
>>>>>
>>>>> I use Aluminum foil to ward off the aliens stealing my thoughts, and
>>>>> of course to ward off WiFi and Cellphone signals... So yeah, that
>>>>> should help with the HyD detectors. ;-)
>>>>>
>>>>> Pete
>>>>>
>>>>> --
>>>>> Peter Gabriel Pitrone - TechRMS
>>>>> Microscopy/Imaging Specialist
>>>>> Prof. Dr. Pavel Tomancak group
>>>>> Max Planck Institute for
>>>>> Molecular Biology and Genetics
>>>>> Pfotenhauerstr. 108
>>>>> 01307 Dresden
>>>>>
>>>>> "If a straight line fit is required, obtain only two data points." -
>>>>> Anon.
>>>>>
>>>>>
>>>>> On Wed, July 10, 2013 8:10 am, Sylvie Le Guyader wrote:
>>>>> <|> *****
>>>>> <|> To join, leave or search the confocal microscopy listserv, go to:
>>>>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>>> <|> *****
>>>>> <|>
>>>>> <|> Hi out there
>>>>> <|>
>>>>> <|> Same experience for us. We were getting random interference
>>>>> lines from  <|> our bright field detector on a NikonA1. It was not
>>>>> linked to cell phones  <|> though but from the cables of another
>>>>> system that stands very nearby.
>>>>> <|> Shielding the cables with alu foil worked a charm.
>>>>> <|> Sylvie
>>>>> <|>
>>>>> <|> On 9 Jul 2013, at 23:50, "Alexander,Steve"
>>>>> <[hidden email]>
>>>>> <|> wrote:
>>>>> <|>
>>>>> <|> *****
>>>>> <|> To join, leave or search the confocal microscopy listserv, go to:
>>>>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>>> <|> *****
>>>>> <|>
>>>>> <|> You could try to shield them with aluminum foil. I tried that to
>>>>> find a  <|> interference source on a different system and it worked!
>>>>> <|>
>>>>> <|> Steve Alexander
>>>>> <|>
>>>>> <|>
>>>>> <|>
>>>>> <|> -----Original Message-----
>>>>> <|> From: Confocal Microscopy List
>>>>> [mailto:[hidden email]]
>>>>> <|> On Behalf Of Yevgeniy Romin
>>>>> <|> Sent: Tuesday, July 09, 2013 4:36 PM  <|> To:
>>>>> [hidden email]
>>>>> <|> Subject: Re: Possible cell phone interference with hybrid
>>>>> detectors  <|>  <|> *****  <|> To join, leave or search the confocal
>>>>> microscopy listserv, go to:
>>>>> <|> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>>> <|> *****
>>>>> <|>
>>>>> <|> Hi Ben
>>>>> <|>
>>>>> <|> We have the same problem with one of our sp5 systems that has
>>>>> HyDs on  it.
>>>>> <|> Every once in a while you see a saturated line across the image.
>>>>> We have  <|> been careful with cell phones around the system, but
>>>>> these lines tend to  <|> happen anyway every once in a while. We had
>>>>> them happen just from a door  <|> opening or closing on the room. My
>>>>> impression is that while cell phones  <|> could be part of the
>>>>> problem, they are probably not the only ones.
>>>>> There
>>>>> <|> could be plenty of equipment or anything else in the room that
>>>>> sometimes  <|> interferes. We have these lines very rarely, but I
>>>>> can't think of a  <|> scenario where they are completely guaranteed
>>>>> not to show up.
>>>>> <|>
>>>>> <|> That being said, we have two other inverted systems, an sp5 with
>>>>> 3 HyDs  <|> and an SP8 with 2 HyDs. We never see any interference
>>>>> lines on these  <|> systems, no matter what is going on in the room.
>>>>> My theory is that the  <|> HyDs on the inverted scanheads are hidden
>>>>> away towards the rear of the  <|> microscope and are perhaps better
>>>>> shielded from interference then those  <|> on the upright system,
>>>>> where they are just sticking out in the open. I  <|> communicated
>>>>> this theory to Leica, and they said that this could  <|> potentially
>>>>> be an explanation, but they haven't yet come up with any  idea  <|>
>>>>> for potential shielding of them.
>>>>> <|>
>>>>> <|> Hope this helps,
>>>>> <|> Yevgeniy
>>>>> <|>
>>>>> <|>
>>>>> <|>
>>>>> <|>
>>>>> <|> Jul 9, 2013, &times; 5:24 PM, "Alison North"
>>>>> <[hidden email]>
>>>>> <|> &Icirc;&Aacute;&ETH;&Eacute;&Oacute;&Aacute;&Igrave;(&Aacute;):
>>>>> <|>
>>>>> <|>> *****
>>>>> <|>> To join, leave or search the confocal microscopy listserv, go to:
>>>>> <|>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>>> <|>> *****
>>>>> <|>>
>>>>> <|>> Hi Ben,
>>>>> <|>>
>>>>> <|>> We were warned by the Leica guys not to use cellphones anywhere
>>>>> near  <|>> the scan head on the SP8 - in fact it has a great big
>>>>> picture of a  <|>> cellphone with a line through it on the scanhead!
>>>>> - so I guess it is a  <|>> well-known problem, not a special "feature"
>>>>> of your system.
>>>>> <|>>
>>>>> <|>> Best wishes,
>>>>> <|>> Alison
>>>>> <|>>
>>>>> <|>> On 7/9/2013 5:13 PM, Smith, Benjamin E. wrote:
>>>>> <|>>> *****
>>>>> <|>>> To join, leave or search the confocal microscopy listserv, go to:
>>>>> <|>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>>>> <|>>> *****
>>>>> <|>>>
>>>>> <|>>> Hi all,
>>>>> <|>>> Our Leica SP8 has a pair of HyD detectors, and I've noticed
>>>>> <|>>> recently that we will get a series of saturated lines during a
>>>>> scan  <|>>> at random, intermittent moments. Judging by their
>>>>> appearance, my  <|>>> guess has been that the detectors were picking
>>>>> up cellular data  <|>>> transmissions. Every time I saw them,
>>>>> someone had a cell phone on  <|>>> them, and having them move the
>>>>> cell phone to the far side of the  <|>>> room, or turning the cell
>>>>> phone off, stopped the lines from  <|>>> occurring.
>>>>> <|>>>
>>>>> <|>>> I was wondering if anyone else has noticed this with hybrid
>>>>> <|>>> detectors, and whether this is normal behavior or something we
>>>>> need  <|>>> to have looked into.
>>>>> <|>>>
>>>>> <|>>> Thanks,
>>>>> <|>>> Ben Smith
>>>>> <|>>>
>>>>> <|>>> Benjamin E. Smith, Ph.D.
>>>>> <|>>> Samuel Roberts Noble Microscopy Laboratory Research Scientist
>>>>> II  <|>>> University of Oklahoma Norman, OK 73019  <|>>> E-mail:
>>>>> [hidden email]  <|>>> Voice 405-325-4391  <|>>> FAX
>>>>> 405-325-7619  <|>>> http://www.microscopy.ou.edu/  <|>>  <|>> --
>>>>> <|>> Alison J. North, Ph.D.,  <|>> Research Associate Professor and
>>>>> <|>> Senior Director of the Bio-Imaging Resource Center, The
>>>>> Rockefeller <|>> University,  <|>> 1230 York Avenue,  <|>> New York,
>>>>> <|>> NY 10065.
>>>>> <|>> Tel: office ++ 212 327 7488
>>>>> <|>> Tel: lab ++ 212 327 7486
>>>>> <|>> Fax: ++ 212 327 7489
>>>>> <|>
>>>>> <|>
>>>>> <|>
>>>>> ====================================================================
>>>>> =
>>>>> <|>
>>>>> <|> Please note that this e-mail and any files transmitted from  <|>
>>>>> Memorial Sloan-Kettering Cancer Center may be privileged,  <|>
>>>>> confidential,  <|> and protected from disclosure under applicable law.
>>>>> If the reader of  <|> this message is not the intended recipient, or
>>>>> an employee or agent  <|> responsible for delivering this message to
>>>>> the intended recipient,  <|> you are hereby notified that any
>>>>> reading, dissemination,  <|> distribution,  <|> copying, or other
>>>>> use of this communication or any of its  attachments  <|> is strictly
>> prohibited.
>>>>> If you have received this communication in  <|> error, please notify
>>>>> the sender immediately by replying to this  <|> message  <|> and
>>>>> deleting this message, any attachments, and all copies and  <|>
>>>>> backups  <|> from your computer.
>>>>> <|>
>> The Babraham Institute, Babraham Research Campus, Cambridge CB22 3AT
>> Registered Charity No. 1053902.
>> The information transmitted in this email is directed only to the
>> addressee. If you received this in error, please contact the sender and
>> delete this email from your system. The contents of this e-mail are the
>> views of the sender and do not necessarily represent the views of the
>> Babraham Institute. Full conditions at: www.babraham.ac.uk<
>> http://www.babraham.ac.uk/email_disclaimer.html>
>>
12