Hello,
I have taken photos using fluorescent beads and am using ImageJ to compare the fluorescence in the sample to the beads. I have been looking at others who have presented semi-quantitative data but there is a huge variation in how it was presented. I was wondering if anyone has done this kind of thing before and what the best way is to present the data (graph, table, etc.). Thanks, Maria Mazzillo Auburn University |
|
|
Maria - I use a fluorescent field made of a concentrated (10% or so) dye. Then you simply divide your image by the image of a dye and get the intensities. I think it's easier than to measure beads.
Mike Model Michael Model, Ph.D. Confocal Microscopy, Dpt Biological Sciences, 1275 University Esplanade, Kent State University, Kent, OH 44242 tel. 330-672-2874 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Maria Mazzillo Sent: Thursday, December 04, 2008 3:40 PM To: [hidden email] Subject: Presenting Semi-quantitative Data Hello, I have taken photos using fluorescent beads and am using ImageJ to compare the fluorescence in the sample to the beads. I have been looking at others who have presented semi-quantitative data but there is a huge variation in how it was presented. I was wondering if anyone has done this kind of thing before and what the best way is to present the data (graph, table, etc.). Thanks, Maria Mazzillo Auburn University |
|
|
I usually use those fluorescent plastic slides which Chroma
give away at conferences. They seem at least as reproducible as a dye solution and much more convenient. There is a range of colours so you can pick one which more or less matches the fluorochrome you are using. Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Electron Microscope Unit, Madsen Building F09, University of Sydney, NSW 2006 ______________________________________________ Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of MODEL, MICHAEL Sent: Friday, 5 December 2008 8:49 AM To: [hidden email] Subject: Re: Presenting Semi-quantitative Data Maria - I use a fluorescent field made of a concentrated (10% or so) dye. Then you simply divide your image by the image of a dye and get the intensities. I think it's easier than to measure beads. Mike Model Michael Model, Ph.D. Confocal Microscopy, Dpt Biological Sciences, 1275 University Esplanade, Kent State University, Kent, OH 44242 tel. 330-672-2874 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Maria Mazzillo Sent: Thursday, December 04, 2008 3:40 PM To: [hidden email] Subject: Presenting Semi-quantitative Data Hello, I have taken photos using fluorescent beads and am using ImageJ to compare the fluorescence in the sample to the beads. I have been looking at others who have presented semi-quantitative data but there is a huge variation in how it was presented. I was wondering if anyone has done this kind of thing before and what the best way is to present the data (graph, table, etc.). Thanks, Maria Mazzillo Auburn University |
|
Peter Zoon |
|
|
In reply to this post by Maria Mazzillo
If you want to do quantitative analysis have a look at
http://www.sipchart.org/. This method uses coverslides with a carefully calibrated dye doped thin polymerfilm to get all kinds of quantitative data from you microscope setup (z-resolution, intensity distribution in xy plane etc. Peter Maria Mazzillo wrote: > Hello, > > I have taken photos using fluorescent beads and am using ImageJ to compare > the fluorescence in the sample to the beads. I have been looking at others > who have presented semi-quantitative data but there is a huge variation in > how it was presented. I was wondering if anyone has done this kind of thing > before and what the best way is to present the data (graph, table, etc.). > > Thanks, > > Maria Mazzillo > Auburn University > > -- Peter D. Zoon Centre for Advanced Microscopy Section of Molecular Cytology Swammerdam Institute for Life Sciences Faculty of Science, University of Amsterdam Visit.mail: Kruislaan 316 (room 2.03) 1098 SM Amsterdam The Netherlands E-mail: [hidden email] Tel: +31-(0)20-5257860 Web: http://wwwmc.bio.uva.nl/ http://www.science.uva.nl/sils http://z00n.net/ |