|
Konstantín Levitskiy |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear all. One of our groups is interested in doing fluorescence ratiometric measurements (e.g., FURA-2, ROS) using a multiphoton microscopy system. From what we understand the laser tuning speeds wont be fast enough to switch between two different excitation wavelengths, and we were wondering if there could be ways to maintain these ratiometric attributes while still performing a laser two-photon excitation. Any ideas or suggestions will be very much appreciated. Thanks and regard, Dr. Konstantín Levitskiy Servicio de Microscopía InstitutodeBiomedicinadeSevilla - IBiS Campus del Hospital Universitario Virgen del Rocío Avda. Manuel Siurot s/nº 41013 Sevilla Tlfno: 955 92 3030 Email: [hidden email] Web: www.ibis-sevilla.es AVISO SOBRE CONFIDENCIALIDAD. Este mensaje y sus anexos son confidenciales y pueden estar protegidos por disposiciones legales. Si Vd. no es el destinatario del mismo, por favor, notifíquenoslo inmediatamente y destruya o devuelva el original. No deberá copiar este mensaje ni sus anexos o usarlo para propósito alguno, ni divulgar su contenido a ninguna persona. P El medio ambiente es nuestra responsabilidad. Por favor, antes de imprimir este mensaje, compruebe que sea necesario hacerlo. |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Konstantin, I would look probes that have common excitation and a shift in emission (Indo-1 for example) Mike Model ________________________________________ From: Confocal Microscopy List <[hidden email]> on behalf of Microscopia-IBIS <[hidden email]> Sent: Monday, April 25, 2016 5:32 AM To: [hidden email] Subject: Ratiometric measurements for multiphoton microscopy ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear all. One of our groups is interested in doing fluorescence ratiometric measurements (e.g., FURA-2, ROS) using a multiphoton microscopy system. From what we understand the laser tuning speeds won’t be fast enough to switch between two different excitation wavelengths, and we were wondering if there could be ways to maintain these ratiometric attributes while still performing a laser two-photon excitation. Any ideas or suggestions will be very much appreciated. Thanks and regard, Dr. Konstantín Levitskiy Servicio de Microscopía InstitutodeBiomedicinadeSevilla - IBiS Campus del Hospital Universitario Virgen del Rocío Avda. Manuel Siurot s/nº 41013 Sevilla Tlfno: 955 92 3030 Email: [hidden email] Web: www.ibis-sevilla.es AVISO SOBRE CONFIDENCIALIDAD. Este mensaje y sus anexos son confidenciales y pueden estar protegidos por disposiciones legales. Si Vd. no es el destinatario del mismo, por favor, notifíquenoslo inmediatamente y destruya o devuelva el original. No deberá copiar este mensaje ni sus anexos o usarlo para propósito alguno, ni divulgar su contenido a ninguna persona. P El medio ambiente es nuestra responsabilidad. Por favor, antes de imprimir este mensaje, compruebe que sea necesario hacerlo. |
|
Johannes Helm |
|
|
In reply to this post by Konstantín Levitskiy
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear Konstantin, we had done something like the thing you describe for a CLSM in the 90s. Please, check these papers: A) P.J. Helm, O. Franksson, and K. Carlsson (1995), A confocal scanning laser microscope for quantitative ratiometric 3D measurements of [Ca2+] and Ca2+ diffusions in living cells stained with Fura-2, Pflügers Archiv - European Journal of Physiology, Vol. 429, pp. 672-681 B) PJ Helm, A Patwardhan, and EMM Manders (1997), A study of the precision of confocal, ratiometric, Fura-2 based [Ca2+] measurements, Cell Calcium 22(4):287-298 These systems might work even in the multi photon case. HOWEVER: Concerning A) 1) Using excitation ratiometric dyes will be troublesome in the multi photon case: a) It is difficult, often impossible, to avoid cross excitation in non-linear excitation. An exception is the CFP-YFP-FRET based Chameleont dye. However, you would need TWO lasers for this dye, which would be rather expensive and space consuming ("stapling" lasers is certainy not recommended, however, you could install them on a sub-table-top shelf, which some Companies selling and producing Optical tables offer). b) Whe using the Chameleont dye, you would need a FRET detection. You might check another paper with a proposal on how to measure FRET: Helm P.J. (2012), Proposal of a New Method for Measuring Foerster Resonance Energy Transfer (FRET) Rapidly, Quantitatively and Non-Destructively, International Journal of Molecular Sciences 13(10):12367-12382 c) Using an emission ratiometic dye might be technically easier, although I know that one needs "this dye" and cannot use "that one" for some reasons. 2) You would have to carefully select your EOMs (or AOMs)! Concerning B) WARNING: One needs a really large number of detected photons in order to be able to use the raw data as basis for a fully quantitative ratiometric analysis. Please, note that Kjell Carlsson and Anders Liljeborg had continued to work on the IMS method, which is a beautiful technology, indeed. It can be used - and has been so, although not for ratiometric imaging, for which it nevertheless would be suitable - for suppressing cross talk and bleed through between different imaging channels. I also allow myself to suggest this, in my opinion outstanding, paper, Carlsson K1, Liljeborg A. Simultaneous confocal lifetime imaging of multiple fluorophores using the intensity-modulated multiple-wavelength scanning (IMS) technique. J Microsc. 1998 Aug;191(2):119-127. Best wishes, Johannes -- P. Johannes Helm, M.Sc. PhD Seniorengineer University of Oslo Institute of Basic Medical Sciences Department of Molecular Medicine Division of Physiology P. O. Box 1103 - Blindern NO-0317 Oslo Norway Voice: +47 228 51159 Fax: +47 228 51278 On 2016-04-25 11:32, Microscopia-IBIS wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your > posting. > ***** > > Dear all. > > > > One of our groups is interested in doing fluorescence ratiometric > measurements (e.g., FURA-2, ROS) using a multiphoton microscopy system. > From > what we understand the laser tuning speeds won't be fast enough to > switch > between two different excitation wavelengths, and we were wondering if > there > could be ways to maintain these ratiometric attributes while still > performing a laser two-photon excitation. Any ideas or suggestions will > be > very much appreciated. > > > > Thanks and regard, > > Dr. Konstantín Levitskiy > > Servicio de Microscopía > > InstitutodeBiomedicinadeSevilla - IBiS > > Campus del Hospital Universitario Virgen del Rocío > > Avda. Manuel Siurot s/nº > > 41013 Sevilla > > Tlfno: 955 92 3030 > > Email: [hidden email] > > Web: www.ibis-sevilla.es > > > > > > AVISO SOBRE CONFIDENCIALIDAD. Este mensaje y sus anexos son > confidenciales y > pueden estar protegidos por disposiciones legales. Si Vd. no es el > destinatario del mismo, por favor, notifíquenoslo inmediatamente y > destruya > o devuelva el original. No deberá copiar este mensaje ni sus anexos o > usarlo > para propósito alguno, ni divulgar su contenido a ninguna persona. > > P El medio ambiente es nuestra responsabilidad. Por favor, antes de > imprimir > este mensaje, compruebe que sea necesario hacerlo. WWW: folk.uio.no/jhelm |
|
Christian Wilms |
|
|
In reply to this post by Konstantín Levitskiy
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Ratiometric measurements using two-photon excitation is not easy. It helps to keep that in mind. Fura-2 *might* work if you use a laser that allow fast enough wavelength switching or use two lasers. But due to the broadened excitation bands for the separate peaks (when compared to one-photon excitation), it is by no means clear that it will be successful. Emission-ratiometric indicators would be the way to go. But the calcium-bound form of Indo-1 has an extremely poor two-photon cross-section, making it unsuited for ratiometric imaging when using 2P-excitation. FRET-based genetically encoded calcium indicators such as YC-2.60 can be used for ratiometric detection using 2P-excitation, but one needs to keep in mind that their calcium binding properties tend to be very non-linear, making calibration difficult. An approach many people have been using, although mostly without the step to actually attempting to quantify the [Ca2+] is to use a mix of dyes: one calcium indicator (e.g. Fluo-4) and one [Ca2+]-insensitive dye (e.g. Alexa 594). This combination yields ratiometric traces with a very good signal-to-noise ratio, but I would be very sceptical of papers trying to actually claim it reflects an absolute calcium concentration. If actually quantifying calcium is what you want to achieve, there are approaches using single wavelength indicators such as Fluo-4 or OGB-1 that might be worth looking into. If you are working in a system that allows you to saturate the indicator in the cell at the end of an experiment, have a look at Miquel Maravall's 2000 Biophysical Journal paper (DOI: 10.1016/S0006-3495(00)76809-3). And finally, a shameless plug: fluorescence lifetime imaging enables ratiometric analysis of a single wavelength indicator. See Wilms et al. Cell Calcium, 2006 (DOI: 10.1016/j.ceca.2006.03.006) Good luck with your experiments! |
|
Christian Wilms |
|
|
In reply to this post by Konstantín Levitskiy
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Ratiometric measurements using two-photon excitation is not easy. It helps to keep that in mind. Fura-2 *might* work if you use a laser that allow fast enough wavelength switching or use two lasers. But due to the broadened excitation bands for the separate peaks (when compared to one-photon excitation), it is by no means clear that it will be successful. Emission-ratiometric indicators would be the way to go. But the calcium-bound form of Indo-1 has an extremely poor two-photon cross-section, making it unsuited for ratiometric imaging when using 2P-excitation. FRET-based genetically encoded calcium indicators such as YC-2.60 can be used for ratiometric detection using 2P-excitation, but one needs to keep in mind that their calcium binding properties tend to be very non-linear, making calibration difficult. An approach many people have been using, although mostly without the step to actually attempting to quantify the [Ca2+] is to use a mix of dyes: one calcium indicator (e.g. Fluo-4) and one [Ca2+]-insensitive dye (e.g. Alexa 594). This combination yields ratiometric traces with a very good signal-to-noise ratio, but I would be very sceptical of papers trying to actually claim it reflects an absolute calcium concentration. If actually quantifying calcium is what you want to achieve, there are approaches using single wavelength indicators such as Fluo-4 or OGB-1 that might be worth looking into. If you are working in a system that allows you to saturate the indicator in the cell at the end of an experiment, have a look at Miquel Maravall's 2000 Biophysical Journal paper (DOI: 10.1016/S0006-3495(00)76809-3). And finally, a shameless plug: fluorescence lifetime imaging enables ratiometric analysis of a single wavelength indicator. See Wilms et al. Cell Calcium, 2006 (DOI: 10.1016/j.ceca.2006.03.006) Good luck with your experiments! Best, Christian |
|
Mark Cannell-2 |
|
|
In reply to this post by Christian Wilms
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi All When we used info-1 with 690 nm 2P excitation, I thought that the ratio image was quite reasonable (see Jones, K.T., Soeller, C., Cannell, M.B., 1998. The passage of Ca2+ and fluorescent markers between the sperm and egg after fusion in the mouse. Development 125, 4627–4635). Also, using a second Ca insensitive dye to construct a “ratio" measurement is unwise -the point of the second wavelength is to control for bleaching and dye distribution etc. my 2c Cheers On 26/04/2016, at 10:27 am, Christian Wilms <[hidden email]> wrote: > > Ratiometric measurements using two-photon excitation is not easy. It helps to keep that in mind. > > Fura-2 *might* work if you use a laser that allow fast enough wavelength switching or use two lasers. But due to the broadened excitation bands for the separate peaks (when compared to one-photon excitation), it is by no means clear that it will be successful. > > Emission-ratiometric indicators would be the way to go. But the calcium-bound form of Indo-1 has an extremely poor two-photon cross-section, making it unsuited for ratiometric imaging when using 2P-excitation. > > FRET-based genetically encoded calcium indicators such as YC-2.60 can be used for ratiometric detection using 2P-excitation, but one needs to keep in mind that their calcium binding properties tend to be very non-linear, making calibration difficult. > > An approach many people have been using, although mostly without the step to actually attempting to quantify the [Ca2+] is to use a mix of dyes: one calcium indicator (e.g. Fluo-4) and one [Ca2+]-insensitive dye (e.g. Alexa 594). This combination yields ratiometric traces with a very good signal-to-noise ratio, but I would be very sceptical of papers trying to actually claim it reflects an absolute calcium concentration. > > If actually quantifying calcium is what you want to achieve, there are approaches using single wavelength indicators such as Fluo-4 or OGB-1 that might be worth looking into. If you are working in a system that allows you to saturate the indicator in the cell at the end of an experiment, have a look at Miquel Maravall's 2000 Biophysical Journal paper (DOI: 10.1016/S0006-3495(00)76809-3). And finally, a shameless plug: fluorescence lifetime imaging enables ratiometric analysis of a single wavelength indicator. See Wilms et al. Cell Calcium, 2006 (DOI: 10.1016/j.ceca.2006.03.006) > > Good luck with your experiments! Mark B. Cannell Ph.D. FRSNZ FISHR Professor of Cardiac Cell Biology School of Physiology & Pharmacology Faculty of Biomedical Sciences University of Bristol Bristol BS8 1TD UK [hidden email] |
|
Christian Wilms |
|
|
In reply to this post by Konstantín Levitskiy
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** "When we used info-1 with 690 nm 2P excitation, I thought that the ratio image was quite reasonable (see Jones, K.T., Soeller, C., Cannell, M.B., 1998. The passage of Ca2+ and fluorescent markers between the sperm and egg after fusion in the mouse. Development 125, 4627-4635)." I admit that I never tried it in cells, only cuvettes. But I could reproduce the more than an order of magnitude lower excitation efficiency than e.g. Fura-2 (http://www.drbio.cornell.edu/cross_sections.html) and decided that the required excitation power was not ideal for my experiments on small neuronal compartments. When imaging larger structures at low zoom, that should be much less of a problem. "Also, using a second Ca insensitive dye to construct a "ratio" measurement is unwise -the point of the second wavelength is to control for bleaching and dye distribution etc." I am also not at all a fan of the approach. It comes with numerous problems, starting with different diffusion rates of the two dyes over different bleaching behaviour, to differential scattering of the two emission wavelengths, leading to a depth-dependent gradient in the ratio, that would need to be corrected for. Having said all that, it has become pretty much standard in neuroscience to report 'green over red' ratios. See publications by the Sabatini for examples. Best, Chris |
|
Mark Cannell-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Chris Any idea what is the cause of the difference between the measurements you cite and the conclusions of Szmacinski et al (Photochem Photobiol. 1993 Sep;58(3):341-5): "The emission spectra and lifetime of Indo-1 appear to be identical for one-photon and two-photon excitation at 351 and 702 mn, respectively, suggesting that the relative one- and two-photon cross sections are similar for the calcium-free and calcium-bound forms of Indo-1. Also, the two-photon cross section of Indo-1 is relatively high, about 4 x 10(-49) cm4 s/photon molecule at 690 nm for both the calcium-free and calcium-bound forms. “? Cheers Mark . Calcium-dependent fluorescence lifetimes of Indo-1 for one- and two-photon excitation of fluorescence. Szmacinski H1, Gryczynski I, Lakowicz JR. On 27/04/2016, at 11:25 am, Christian Wilms <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > "When we used info-1 with 690 nm 2P excitation, I thought that the ratio image was quite reasonable (see Jones, K.T., Soeller, C., Cannell, M.B., 1998. The passage of Ca2+ and fluorescent markers between the sperm and egg after fusion in the mouse. Development 125, 4627-4635)." > > I admit that I never tried it in cells, only cuvettes. But I could reproduce the more than an order of magnitude lower excitation efficiency than e.g. Fura-2 (http://www.drbio.cornell.edu/cross_sections.html) and decided that the required excitation power was not ideal for my experiments on small neuronal compartments. When imaging larger structures at low zoom, that should be much less of a problem. > > "Also, using a second Ca insensitive dye to construct a "ratio" measurement is unwise -the point of the second wavelength is to control for bleaching and dye distribution etc." > > I am also not at all a fan of the approach. It comes with numerous problems, starting with different diffusion rates of the two dyes over different bleaching behaviour, to differential scattering of the two emission wavelengths, leading to a depth-dependent gradient in the ratio, that would need to be corrected for. Having said all that, it has become pretty much standard in neuroscience to report 'green over red' ratios. See publications by the Sabatini for examples. > > Best, Chris Mark B. Cannell Ph.D. FRSNZ FISHR Professor of Cardiac Cell Biology School of Physiology & Pharmacology Faculty of Biomedical Sciences University of Bristol Bristol BS8 1TD UK [hidden email] |
|
Christian Wilms |
|
|
In reply to this post by Konstantín Levitskiy
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Mark, >Any idea what is the cause of the difference between the measurements you cite and the conclusions of Szmacinski et al (Photochem Photobiol. 1993 Sep;58(3):341-5): "The emission spectra and lifetime of Indo-1 appear to be identical for one-photon and two- photon excitation at 351 and 702 mn, respectively, suggesting that the relative one- and two-photon cross sections are similar for the calcium-free and calcium-bound forms of Indo- 1. Also, the two-photon cross section of Indo-1 is relatively high, about 4 x 10(-49) cm4 s/photon molecule at 690 nm for both the calcium-free and calcium-bound forms. “? I can only hazard a guess, as it is surprising that the two reported values are so different. Reading the Szmacinski paper, they seem to have estimated the cross-section by comparison with a standard. In constrast Xu and Webb (http://www.pnas.org/content/93/20/10763.long) seem to have measured it directly. But, as the former estimation is common methodology, I doubt that alone can explain the difference. Best, Chris |
|
De Tombe, Pieter |
|
|
In reply to this post by Christian Wilms
Hi,
We used 2P exception to measure radiometric Indo-1 Ca2+ transients in cardiac myocytes; we needed red excitation to prevent breakdown of Blebbistatin, a compound that uncouples contractile proteins from Ca2+ activation but is photo-inactivated by blue light. see: Farman GP, Tachampa K, Mateja R, Cazorla O, Lacampagne A, de Tombe PP. Blebbistatin: use as inhibitor of muscle contraction. Pflugers Arch. 2008 Mar;455(6):995–1005. ----------------------------------------------------------- Pieter P. de Tombe, Ph.D. James R. DePauw Professor of Physiology Chair, Department of Cell and Molecular Physiology Loyola University Chicago 2160 South First Ave. Stritch School of Medicine Maywood, IL 60153-5500 (708) 216-1018 (708) 216-6308 (FAX) [hidden email] > On Apr 26, 2016, at 5:01 PM, Christian Wilms <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Ratiometric measurements using two-photon excitation is not easy. It helps to keep that in mind. > > Fura-2 *might* work if you use a laser that allow fast enough wavelength switching or use two lasers. But due to the broadened excitation bands for the separate peaks (when compared to one-photon excitation), it is by no means clear that it will be successful. > > Emission-ratiometric indicators would be the way to go. But the calcium-bound form of Indo-1 has an extremely poor two-photon cross-section, making it unsuited for ratiometric imaging when using 2P-excitation. > > FRET-based genetically encoded calcium indicators such as YC-2.60 can be used for ratiometric detection using 2P-excitation, but one needs to keep in mind that their calcium binding properties tend to be very non-linear, making calibration difficult. > > An approach many people have been using, although mostly without the step to actually attempting to quantify the [Ca2+] is to use a mix of dyes: one calcium indicator (e.g. Fluo-4) and one [Ca2+]-insensitive dye (e.g. Alexa 594). This combination yields ratiometric traces with a very good signal-to-noise ratio, but I would be very sceptical of papers trying to actually claim it reflects an absolute calcium concentration. > > If actually quantifying calcium is what you want to achieve, there are approaches using single wavelength indicators such as Fluo-4 or OGB-1 that might be worth looking into. If you are working in a system that allows you to saturate the indicator in the cell at the end of an experiment, have a look at Miquel Maravall's 2000 Biophysical Journal paper (DOI: 10.1016/S0006-3495(00)76809-3). And finally, a shameless plug: fluorescence lifetime imaging enables ratiometric analysis of a single wavelength indicator. See Wilms et al. Cell Calcium, 2006 (DOI: 10.1016/j.ceca.2006.03.006) > > Good luck with your experiments! > > Best, Christian |
|
Christian Wilms |
|
|
In reply to this post by Konstantín Levitskiy
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** My apologies for the previous message. It seems to have been caught in the system for nearly 48 hours. Until now that is. |
|
De Tombe, Pieter |
|
|
In reply to this post by De Tombe, Pieter
Oops, auto corrector mishap ... 2P excitation … sorry
P. > On Apr 28, 2016, at 9:37 AM, De Tombe, Pieter <[hidden email]> wrote: > > Hi, > > We used 2P exception to measure radiometric Indo-1 Ca2+ transients in cardiac myocytes; we needed red excitation to prevent breakdown of Blebbistatin, a compound that uncouples > contractile proteins from Ca2+ activation but is photo-inactivated by blue light. > see: Farman GP, Tachampa K, Mateja R, Cazorla O, Lacampagne A, de Tombe PP. Blebbistatin: use as inhibitor of muscle contraction. Pflugers Arch. 2008 Mar;455(6):995–1005. > > ----------------------------------------------------------- > Pieter P. de Tombe, Ph.D. > James R. DePauw Professor of Physiology > Chair, Department of Cell and Molecular Physiology > Loyola University Chicago > 2160 South First Ave. > Stritch School of Medicine > Maywood, IL 60153-5500 > (708) 216-1018 > (708) 216-6308 (FAX) > [hidden email] > >> On Apr 26, 2016, at 5:01 PM, Christian Wilms <[hidden email]> wrote: >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Ratiometric measurements using two-photon excitation is not easy. It helps to keep that in mind. >> >> Fura-2 *might* work if you use a laser that allow fast enough wavelength switching or use two lasers. But due to the broadened excitation bands for the separate peaks (when compared to one-photon excitation), it is by no means clear that it will be successful. >> >> Emission-ratiometric indicators would be the way to go. But the calcium-bound form of Indo-1 has an extremely poor two-photon cross-section, making it unsuited for ratiometric imaging when using 2P-excitation. >> >> FRET-based genetically encoded calcium indicators such as YC-2.60 can be used for ratiometric detection using 2P-excitation, but one needs to keep in mind that their calcium binding properties tend to be very non-linear, making calibration difficult. >> >> An approach many people have been using, although mostly without the step to actually attempting to quantify the [Ca2+] is to use a mix of dyes: one calcium indicator (e.g. Fluo-4) and one [Ca2+]-insensitive dye (e.g. Alexa 594). This combination yields ratiometric traces with a very good signal-to-noise ratio, but I would be very sceptical of papers trying to actually claim it reflects an absolute calcium concentration. >> >> If actually quantifying calcium is what you want to achieve, there are approaches using single wavelength indicators such as Fluo-4 or OGB-1 that might be worth looking into. If you are working in a system that allows you to saturate the indicator in the cell at the end of an experiment, have a look at Miquel Maravall's 2000 Biophysical Journal paper (DOI: 10.1016/S0006-3495(00)76809-3). And finally, a shameless plug: fluorescence lifetime imaging enables ratiometric analysis of a single wavelength indicator. See Wilms et al. Cell Calcium, 2006 (DOI: 10.1016/j.ceca.2006.03.006) >> >> Good luck with your experiments! >> >> Best, Christian > |
| Free forum by Nabble | Edit this page |