|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** *****commercial reply ***** Hi all, we wanted to add some information to the discussion about the CPU, RAM, Disk and Graphics card. We cannot speak for acquisition software requirements but only for analysis software needs. All that Oliver Biehlmaier said about CPU, RAM etc. applies also for most cases in Imaris. If the data and derived data fit into the RAM (rule of thumb ten times the amount of RAM of the raw data set) the CPU speed will be the limiting factor as well as the core count. Imaris is in many parts multithreaded since Version 6.3 from IO operations to image filters. But you can still get the impression that it uses only one or two cores as some parts are not multi-threaded which may dominate the execution time. We continue to make additional functions multithreaded like in the latest version the distance transformations and diameter calculation in Filament Tracer. However, to come back to Olivers post; our current hardware recommendation for Imaris in general is: - get a fast 4 (6) core processor (with hyper-threading). This is currently the sweet spot in price-performance ratio; Imaris is making good use of this 8 or 12 threads; - enough RAM; about 10x of the raw data set size to fit the data set and its analysis into RAM; - if possible get a fast SSD only used as Imaris cache drive (for cases where the data set would not fit into the RAM); at least 10x the size of your typical raw data set size; - Get a high-end gaming graphic card (nVidia or ATI); unless you want to do stereo 3D with shutter glasses in which case nVidia Quadro card is required. Our take home message is: keep your CPU fed with data as fast as possible! EU Support Team Bitplane AG |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear all, this seems not to be a proper confocal question but I know your great expertise could help me in solve this as the IT couldn't..... I am trying to recover the informations of logins on 2 computers in our facility ( to compare the real usage of the system). Both computers run on Windows7 and the Event viewer I can visualize do not contain the information on the name of the user logged in. As for Login user the system gave to me the name of the computer and a $ close to that. I managed so far to have the Event viewer working on all the Windows XP systems, but it seems Windows7 works in a different way. Thanks in advance for all your useful replies. Valeria -- Valeria Berno,PhD Microscopy Facility Manager MRC Centre for Regenerative Medicine SCRM building The University of Edinburgh Edinburgh bioQuarter 5 Little France Drive Edinburgh EH16 4UU Tel Office 0131 6519521 The University of Edinburgh is a charitable body, registered in Scotland, with registration number SC005336. |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Valeria, The windows 7 event viewer is different than XPs viewer. I have been unable to turn on a column to show users. However, the users name can be found in the Details tab of the Security log. Check on Logon events, some of them near the end of the logon process will have the users name. I find that it is easier to look at Logoff entries, as it is easier to find the users name there, then I go back to find the time of the login. I hope this helps, Lon -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Valeria Berno Sent: Wednesday, March 20, 2013 10:47 AM To: [hidden email] Subject: Login events on Windows 7 ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear all, this seems not to be a proper confocal question but I know your great expertise could help me in solve this as the IT couldn't..... I am trying to recover the informations of logins on 2 computers in our facility ( to compare the real usage of the system). Both computers run on Windows7 and the Event viewer I can visualize do not contain the information on the name of the user logged in. As for Login user the system gave to me the name of the computer and a $ close to that. I managed so far to have the Event viewer working on all the Windows XP systems, but it seems Windows7 works in a different way. Thanks in advance for all your useful replies. Valeria -- Valeria Berno,PhD Microscopy Facility Manager MRC Centre for Regenerative Medicine SCRM building The University of Edinburgh Edinburgh bioQuarter 5 Little France Drive Edinburgh EH16 4UU Tel Office 0131 6519521 The University of Edinburgh is a charitable body, registered in Scotland, with registration number SC005336. |
|
|
In reply to this post by Valeria Berno-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Valeria, Do you know Agendo (cirklo.org/wiki)? It's an open source resource manager that enables you to do so. Please contact me off list if interested on further details. Best, Nuno Moreno, PhD Instituto Gulbenkian de Ciência On Mar 20, 2013, at 3:46 PM, Valeria Berno wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all, > > this seems not to be a proper confocal question but I know your great expertise could help me in solve this as the IT couldn't..... > > I am trying to recover the informations of logins on 2 computers in our facility ( to compare the real usage of the system). Both computers run on Windows7 and the Event viewer I can visualize do not contain the information on the name of the user logged in. As for Login user the system gave to me the name of the computer and a $ close to that. > > I managed so far to have the Event viewer working on all the Windows XP systems, but it seems Windows7 works in a different way. > > Thanks in advance for all your useful replies. > > Valeria > > -- > Valeria Berno,PhD > Microscopy Facility Manager > MRC Centre for Regenerative Medicine > SCRM building > The University of Edinburgh > Edinburgh bioQuarter > 5 Little France Drive > Edinburgh > EH16 4UU > Tel Office 0131 6519521 > > > The University of Edinburgh is a charitable body, registered in > Scotland, with registration number SC005336. |
|
Graham Wright-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello Valeria, For logons only I've used the guidance here in the past: http://msmvps.com/blogs/kwsupport/archive/2005/02/24/36942.aspx to create a log locally on each of the microscope computers by setting the location to save as D:\directory\log\log_on_off.log (for example). It works nicely if you just need a reference for who has been using you microscopes. My computer guys then helped do this at the server level (Samba; logon.bat) running a logon script for domain users, which looks like: ----------------- echo Login From: %COMPUTERNAME%, User Name: %USERNAME%, Date: %DATE%, Time: %TIME% >> \\server\directory\logs\logons.log ----------------- So, every users that log into the domain will have the logon information captured. As an example the log file will look like below and I can open it on any computer with access to the server. I can then search for the computer name I want to know about (as we have about 20 computers, mostly running XP) ------------------ Login From: SELETAR, User Name: darren, Date: Fri 01/11/2013, Time: 19:01:47.62 Login From: PALAWAN, User Name: yuezhen, Date: Sat 01/12/2013, Time: 9:54:44.65 ------------------ This works well as no configurations need to be done on the windows domain clients. To capture log off information has proved trickier and we are still trying to find the best workaround for that. Fortunately I don't need that information at present. Regards, Graham -- Dr Graham Wright Microscopy Unit Manager Institute of Medical Biology 8A Biomedical Grove, #06-06 Immunos, Singapore 138648 O: +65 6407 0167 E: [hidden email] W: http://www.imb.a-star.edu.sg/imu/ On 21 March 2013 07:19, Nuno Moreno <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Valeria, > > Do you know Agendo (cirklo.org/wiki)? It's an open source resource > manager that enables you to do so. Please contact me off list if interested > on further details. > > Best, > Nuno Moreno, PhD > Instituto Gulbenkian de Ciência > > > > > > On Mar 20, 2013, at 3:46 PM, Valeria Berno wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Dear all, > > > > this seems not to be a proper confocal question but I know your great > expertise could help me in solve this as the IT couldn't..... > > > > I am trying to recover the informations of logins on 2 computers in our > facility ( to compare the real usage of the system). Both computers run on > Windows7 and the Event viewer I can visualize do not contain the > information on the name of the user logged in. As for Login user the system > gave to me the name of the computer and a $ close to that. > > > > I managed so far to have the Event viewer working on all the Windows XP > systems, but it seems Windows7 works in a different way. > > > > Thanks in advance for all your useful replies. > > > > Valeria > > > > -- > > Valeria Berno,PhD > > Microscopy Facility Manager > > MRC Centre for Regenerative Medicine > > SCRM building > > The University of Edinburgh > > Edinburgh bioQuarter > > 5 Little France Drive > > Edinburgh > > EH16 4UU > > Tel Office 0131 6519521 > > > > > > The University of Edinburgh is a charitable body, registered in > > Scotland, with registration number SC005336. > |
|
|
In reply to this post by Valeria Berno-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I had previously written some AWK scripts to pull out all the needed logon/logoff information from a "Save Log File As…" export of the Security Log Events file from our Windows XP computers, but as noted in the post from March, Windows 7 Logfiles have very different structures, so the AWK scripts no longer work. However, I have very recently been testing the PowerShell script at: http://gallery.technet.microsoft.com/scriptcenter/Log-Parser-to-Identify-8aac36bd (FYI Description of security events in Windows 7 at http://support.microsoft.com/kb/977519) It seems to work quite well on Windows 7 (and probably on Windows XP as well), and when you choose the "Table" option when you run the script in PowerShell, it produces an onscreen table of all Logon and Logoff events (and not all the system and network logon/logoff events) that you can select and paste into a document; alternately, you can modify the script to produce a file output automatically when you run the script - and no need for any other additional AWK or other scripts. Sorry for the long delay since the original post back in March, but have just been testing this recently and hope that this PowerShell script might still be useful to people. --Bruce ************************************************* Bruce A. Stanley, Ph.D. Director, Scientific Programs Section of Research Resources H093, Room C1734 Director, Mass Spectrometry and Proteomics Facility Co-Director, CTSI Translational Technologies Core Services Unit (TTCSU) Penn State College of Medicine 500 University Drive Hershey, PA 17033-2390 Office/Lab: (717) 531-5329 FAX: (717) 531-0239 Email: [hidden email] WEB page: http://med.psu.edu/web/core/bruce-stanley-research ************************************************* *****E-Mail Confidentiality Notice***** This message (including any attachments) contains information intended for a specific individual(s) and purpose that may be privileged, confidential or otherwise protected from disclosure pursuant to applicable law. Any inappropriate use, distribution or copying of the message is strictly prohibited and may subject you to criminal or civil penalty. If you have received this transmission in error, please reply to the sender indicating this error and delete the transmission from your system immediately. On Mar 20, 2013, at 11:46 AM, Valeria Berno <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all, > > this seems not to be a proper confocal question but I know your great expertise could help me in solve this as the IT couldn't..... > > I am trying to recover the informations of logins on 2 computers in our facility ( to compare the real usage of the system). Both computers run on Windows7 and the Event viewer I can visualize do not contain the information on the name of the user logged in. As for Login user the system gave to me the name of the computer and a $ close to that. > > I managed so far to have the Event viewer working on all the Windows XP systems, but it seems Windows7 works in a different way. > > Thanks in advance for all your useful replies. > > Valeria > > -- > Valeria Berno,PhD > Microscopy Facility Manager > MRC Centre for Regenerative Medicine > SCRM building > The University of Edinburgh > Edinburgh bioQuarter > 5 Little France Drive > Edinburgh > EH16 4UU > Tel Office 0131 6519521 > > > The University of Edinburgh is a charitable body, registered in > Scotland, with registration number SC005336. |
| Free forum by Nabble | Edit this page |