Re: Objective lens chromatic aberration

classic Classic list List threaded Threaded
♦
♦
16 messages Options
mcammer mcammer
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Objective lens chromatic aberration

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I originally posted the message below to the microscopy listserv but reviewing the discussion here on chromatic aberration, I thought it would be apropos here too even though it is not really regarding confocal.

A few years ago we were having problems with the first commercial Olympus TIRF system because we could not get consistent evanescent waves with the one angle adjustment with the laser lines we had from 405 to 568 nm that were delivered via a single fiber (it was worse when we later added a 633 nm laser).  I suggested we pump each laser in through a separate path that could be angled independently.  We didn't build it, but I think Olympus now sells a TIRF system that does this.

Another issue is that when I first heard about TIRF maybe 15 years ago, it was introduced as a ring illumination at the outer edge of the back aperture, not as a single point or crescent at the periphery on only one side.  A ring, or at least a series of points around the periphery, seems like a better way to provide a uniform field due to aberrations from coherent light in the imperfect optics.  Any thought on this?

Sincerely,

Michael

-----------------------ORIGINAL MESSAGE-------------------------------
We have the Nikon TIRF system and have three laser lines
going into the TIRF arm via a single fiber.  When we project through
the 100X objective through the sample onto the wall we see that the
lines go through the sample at different angles.  (You can see a
picture of the projection at approx 45 degrees at
http://www.flickr.com/photos/mcammer/5359189090/ .)  It is also
noticeable in the TIRF images that the field depth is different for
each wavelength.  Is this unavoidable due to the different
wavelengths or is it possible to align the optics better so these
spots would be more coincident?



_________________________________________
Michael Cammer, Assistant Research Scientist
Skirball Institute of Biomolecular Medicine
Lab: (212) 263-3208  Cell: (914) 309-3270


------------------------------------------------------------
This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email.
=================================
Gary G. Li-2 Gary G. Li-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Objective lens chromatic aberration

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Michael,

There are a few factors that need to be compromised in TIRF.   1) The TIRF
penetration depth is dependent upon the wavelength of the incident
illumination, the angle of incidence, and the refractive indices of the
media at the interface. 2) An apochromatic lens brings lasers of three
wavelengths on to the same focal plane, but may not to the same incident
angle.  Therefore, the incident angles for the three lasers should be able
to be adjusted independently to get optimal TIRF images.

You may want to take your sample off first, to see how different the angles
of the three lasers projecting to the wall, to get an idea of the incident
angles of your lasers coming out of the objective.

Regards

Gary G Li, PhD


On Sun, Jan 16, 2011 at 1:02 PM, Cammer, Michael <[hidden email]
> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> I originally posted the message below to the microscopy listserv but
> reviewing the discussion here on chromatic aberration, I thought it would be
> apropos here too even though it is not really regarding confocal.
>
> A few years ago we were having problems with the first commercial Olympus
> TIRF system because we could not get consistent evanescent waves with the
> one angle adjustment with the laser lines we had from 405 to 568 nm that
> were delivered via a single fiber (it was worse when we later added a 633 nm
> laser).  I suggested we pump each laser in through a separate path that
> could be angled independently.  We didn't build it, but I think Olympus now
> sells a TIRF system that does this.
>
> Another issue is that when I first heard about TIRF maybe 15 years ago, it
> was introduced as a ring illumination at the outer edge of the back
> aperture, not as a single point or crescent at the periphery on only one
> side.  A ring, or at least a series of points around the periphery, seems
> like a better way to provide a uniform field due to aberrations from
> coherent light in the imperfect optics.  Any thought on this?
>
> Sincerely,
>
> Michael
>
> -----------------------ORIGINAL MESSAGE-------------------------------
> We have the Nikon TIRF system and have three laser lines
> going into the TIRF arm via a single fiber.  When we project through
> the 100X objective through the sample onto the wall we see that the
> lines go through the sample at different angles.  (You can see a
> picture of the projection at approx 45 degrees at
> http://www.flickr.com/photos/mcammer/5359189090/ .)  It is also
> noticeable in the TIRF images that the field depth is different for
> each wavelength.  Is this unavoidable due to the different
> wavelengths or is it possible to align the optics better so these
> spots would be more coincident?
>
>
>
> _________________________________________
> Michael Cammer, Assistant Research Scientist
> Skirball Institute of Biomolecular Medicine
> Lab: (212) 263-3208  Cell: (914) 309-3270
>
>
> ------------------------------------------------------------
> This email message, including any attachments, is for the sole use of the
> intended recipient(s) and may contain information that is proprietary,
> confidential, and exempt from disclosure under applicable law. Any
> unauthorized review, use, disclosure, or distribution is prohibited. If you
> have received this email in error please notify the sender by return email
> and delete the original message. Please note, the recipient should check
> this email and any attachments for the presence of viruses. The organization
> accepts no liability for any damage caused by any virus transmitted by this
> email.
> =================================
>
Mark Cannell Mark Cannell
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Objective lens chromatic aberration

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I suggest it is much more likely that the fiber/rear aperture coupling  
lens was at fault.  I must admit I don't see how with rear aperture  
illumination the same and same focal length for the 3 lasers  how the  
incident rays could have different angles ...

Cheers


> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Michael,
>
> There are a few factors that need to be compromised in TIRF.   1)  
> The TIRF
> penetration depth is dependent upon the wavelength of the incident
> illumination, the angle of incidence, and the refractive indices of  
> the
> media at the interface. 2) An apochromatic lens brings lasers of three
> wavelengths on to the same focal plane, but may not to the same  
> incident
> angle.  Therefore, the incident angles for the three lasers should  
> be able
> to be adjusted independently to get optimal TIRF images.
>
> You may want to take your sample off first, to see how different the  
> angles
> of the three lasers projecting to the wall, to get an idea of the  
> incident
> angles of your lasers coming out of the objective.
>
> Regards
>
> Gary G Li, PhD
>
>
> On Sun, Jan 16, 2011 at 1:02 PM, Cammer, Michael <[hidden email]
>> wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> I originally posted the message below to the microscopy listserv but
>> reviewing the discussion here on chromatic aberration, I thought it  
>> would be
>> apropos here too even though it is not really regarding confocal.
>>
>> A few years ago we were having problems with the first commercial  
>> Olympus
>> TIRF system because we could not get consistent evanescent waves  
>> with the
>> one angle adjustment with the laser lines we had from 405 to 568 nm  
>> that
>> were delivered via a single fiber (it was worse when we later added  
>> a 633 nm
>> laser).  I suggested we pump each laser in through a separate path  
>> that
>> could be angled independently.  We didn't build it, but I think  
>> Olympus now
>> sells a TIRF system that does this.
>>
>> Another issue is that when I first heard about TIRF maybe 15 years  
>> ago, it
>> was introduced as a ring illumination at the outer edge of the back
>> aperture, not as a single point or crescent at the periphery on  
>> only one
>> side.  A ring, or at least a series of points around the periphery,  
>> seems
>> like a better way to provide a uniform field due to aberrations from
>> coherent light in the imperfect optics.  Any thought on this?
>>
>> Sincerely,
>>
>> Michael
>>
>> -----------------------ORIGINAL  
>> MESSAGE-------------------------------
>> We have the Nikon TIRF system and have three laser lines
>> going into the TIRF arm via a single fiber.  When we project through
>> the 100X objective through the sample onto the wall we see that the
>> lines go through the sample at different angles.  (You can see a
>> picture of the projection at approx 45 degrees at
>> http://www.flickr.com/photos/mcammer/5359189090/ .)  It is also
>> noticeable in the TIRF images that the field depth is different for
>> each wavelength.  Is this unavoidable due to the different
>> wavelengths or is it possible to align the optics better so these
>> spots would be more coincident?
>>
>>
>>
>> _________________________________________
>> Michael Cammer, Assistant Research Scientist
>> Skirball Institute of Biomolecular Medicine
>> Lab: (212) 263-3208  Cell: (914) 309-3270
>>
>>
>> ------------------------------------------------------------
>> This email message, including any attachments, is for the sole use  
>> of the
>> intended recipient(s) and may contain information that is  
>> proprietary,
>> confidential, and exempt from disclosure under applicable law. Any
>> unauthorized review, use, disclosure, or distribution is  
>> prohibited. If you
>> have received this email in error please notify the sender by  
>> return email
>> and delete the original message. Please note, the recipient should  
>> check
>> this email and any attachments for the presence of viruses. The  
>> organization
>> accepts no liability for any damage caused by any virus transmitted  
>> by this
>> email.
>> =================================
>>
Andreas Bruckbauer Andreas Bruckbauer
Reply | Threaded
Open this post in threaded view
♦
♦
|

TIRF and chromatic aberration

In reply to this post by mcammer
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****


For the point illumination it is just much easier to adjust the angle, i think the Till Photonics system can rotate the point very fast to give the same effect as the ring. An advantage of a ring would be that one does not destroy the coating on the objective when using high laser power.

Olympus had a three laser TIRF illuminator for several years, at least in Europe. This allows to adjust the angle, position and focus for each wavelength independetly and it works very well for simultaenous multi-color TIRF with an image splitter (I have no commercial interest). They now have a four color version which is motorized and also marketed in the US.

The Olympus 150x TIRF objective which we use for single molecule work is an Apochromat but not a PLAN Apo and there is some measurable difference is focus over the field of view. In the center i measured an axial chromatic aberration of only 190 nm for emision at  520 nm and 650 nm .
 
best wishes

Andreas

-----Original Message-----
From: Cammer, Michael <[hidden email]>
To: [hidden email]
Sent: Sun, 16 Jan 2011 18:02
Subject: Re: Objective lens chromatic aberration


*****
o join, leave or search the confocal microscopy listserv, go to:
ttp://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
****
I originally posted the message below to the microscopy listserv but reviewing
he discussion here on chromatic aberration, I thought it would be apropos here
oo even though it is not really regarding confocal.
A few years ago we were having problems with the first commercial Olympus TIRF
ystem because we could not get consistent evanescent waves with the one angle
djustment with the laser lines we had from 405 to 568 nm that were delivered
ia a single fiber (it was worse when we later added a 633 nm laser).  I
uggested we pump each laser in through a separate path that could be angled
ndependently.  We didn't build it, but I think Olympus now sells a TIRF system
hat does this.
Another issue is that when I first heard about TIRF maybe 15 years ago, it was
ntroduced as a ring illumination at the outer edge of the back aperture, not as
 single point or crescent at the periphery on only one side.  A ring, or at
east a series of points around the periphery, seems like a better way to
rovide a uniform field due to aberrations from coherent light in the imperfect
ptics.  Any thought on this?
Sincerely,
Michael
-----------------------ORIGINAL MESSAGE-------------------------------
e have the Nikon TIRF system and have three laser lines
oing into the TIRF arm via a single fiber.  When we project through
he 100X objective through the sample onto the wall we see that the
ines go through the sample at different angles.  (You can see a
icture of the projection at approx 45 degrees at
ttp://www.flickr.com/photos/mcammer/5359189090/ .)  It is also
oticeable in the TIRF images that the field depth is different for
ach wavelength.  Is this unavoidable due to the different
avelengths or is it possible to align the optics better so these
pots would be more coincident?

_________________________________________
ichael Cammer, Assistant Research Scientist
kirball Institute of Biomolecular Medicine
ab: (212) 263-3208  Cell: (914) 309-3270

-----------------------------------------------------------
his email message, including any attachments, is for the sole use of the
ntended recipient(s) and may contain information that is proprietary,
onfidential, and exempt from disclosure under applicable law. Any unauthorized
eview, use, disclosure, or distribution is prohibited. If you have received
his email in error please notify the sender by return email and delete the
riginal message. Please note, the recipient should check this email and any
ttachments for the presence of viruses. The organization accepts no liability
or any damage caused by any virus transmitted by this email.
================================
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Objective lens chromatic aberration

In reply to this post by Mark Cannell
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

This would also lead to chromatically dependent NA, which would be a big
no-no for a corrected objective.

Craig


On Sun, Jan 16, 2011 at 2:16 PM, Mark Cannell <[hidden email]>wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> I suggest it is much more likely that the fiber/rear aperture coupling lens
> was at fault.  I must admit I don't see how with rear aperture illumination
> the same and same focal length for the 3 lasers  how the incident rays could
> have different angles ...
>
> Cheers
>
>
>
>  *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> Hi Michael,
>>
>> There are a few factors that need to be compromised in TIRF.   1) The TIRF
>> penetration depth is dependent upon the wavelength of the incident
>> illumination, the angle of incidence, and the refractive indices of the
>> media at the interface. 2) An apochromatic lens brings lasers of three
>> wavelengths on to the same focal plane, but may not to the same incident
>> angle.  Therefore, the incident angles for the three lasers should be able
>> to be adjusted independently to get optimal TIRF images.
>>
>> You may want to take your sample off first, to see how different the
>> angles
>> of the three lasers projecting to the wall, to get an idea of the incident
>> angles of your lasers coming out of the objective.
>>
>> Regards
>>
>> Gary G Li, PhD
>>
>>
>> On Sun, Jan 16, 2011 at 1:02 PM, Cammer, Michael <
>> [hidden email]
>>
>>> wrote:
>>>
>>
>>  *****
>>> To join, leave or search the confocal microscopy listserv, go to:
>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>>> *****
>>>
>>> I originally posted the message below to the microscopy listserv but
>>> reviewing the discussion here on chromatic aberration, I thought it would
>>> be
>>> apropos here too even though it is not really regarding confocal.
>>>
>>> A few years ago we were having problems with the first commercial Olympus
>>> TIRF system because we could not get consistent evanescent waves with the
>>> one angle adjustment with the laser lines we had from 405 to 568 nm that
>>> were delivered via a single fiber (it was worse when we later added a 633
>>> nm
>>> laser).  I suggested we pump each laser in through a separate path that
>>> could be angled independently.  We didn't build it, but I think Olympus
>>> now
>>> sells a TIRF system that does this.
>>>
>>> Another issue is that when I first heard about TIRF maybe 15 years ago,
>>> it
>>> was introduced as a ring illumination at the outer edge of the back
>>> aperture, not as a single point or crescent at the periphery on only one
>>> side.  A ring, or at least a series of points around the periphery, seems
>>> like a better way to provide a uniform field due to aberrations from
>>> coherent light in the imperfect optics.  Any thought on this?
>>>
>>> Sincerely,
>>>
>>> Michael
>>>
>>> -----------------------ORIGINAL MESSAGE-------------------------------
>>> We have the Nikon TIRF system and have three laser lines
>>> going into the TIRF arm via a single fiber.  When we project through
>>> the 100X objective through the sample onto the wall we see that the
>>> lines go through the sample at different angles.  (You can see a
>>> picture of the projection at approx 45 degrees at
>>> http://www.flickr.com/photos/mcammer/5359189090/ .)  It is also
>>> noticeable in the TIRF images that the field depth is different for
>>> each wavelength.  Is this unavoidable due to the different
>>> wavelengths or is it possible to align the optics better so these
>>> spots would be more coincident?
>>>
>>>
>>>
>>> _________________________________________
>>> Michael Cammer, Assistant Research Scientist
>>> Skirball Institute of Biomolecular Medicine
>>> Lab: (212) 263-3208  Cell: (914) 309-3270
>>>
>>>
>>> ------------------------------------------------------------
>>> This email message, including any attachments, is for the sole use of the
>>> intended recipient(s) and may contain information that is proprietary,
>>> confidential, and exempt from disclosure under applicable law. Any
>>> unauthorized review, use, disclosure, or distribution is prohibited. If
>>> you
>>> have received this email in error please notify the sender by return
>>> email
>>> and delete the original message. Please note, the recipient should check
>>> this email and any attachments for the presence of viruses. The
>>> organization
>>> accepts no liability for any damage caused by any virus transmitted by
>>> this
>>> email.
>>> =================================
>>>
>>>
Andreas Bruckbauer Andreas Bruckbauer
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Objective lens chromatic aberration

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I guess the initial question about chromatic aberration and the new one about TIRF got muddeled up. I guess chromatically different NA in TIRF could arise when part of the correction is build in the tube lens? The illumination light forTIRF "through the objective" does not pass the tube lens and is thus is not well corrected. Furthermore slight imperfections in the dichroic which relfects the laser light can lead to distortions and these can be wavelength dependent.  

 
best wishes

Andreas

 

-----Original Message-----
From: Craig Brideau <[hidden email]>
To: [hidden email]
Sent: Mon, 17 Jan 2011 19:01
Subject: Re: Objective lens chromatic aberration


*****

To join, leave or search the confocal microscopy listserv, go to:

http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

*****



This would also lead to chromatically dependent NA, which would be a big

no-no for a corrected objective.



Craig





On Sun, Jan 16, 2011 at 2:16 PM, Mark Cannell <[hidden email]>wrote:



> *****

> To join, leave or search the confocal microscopy listserv, go to:

> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

> *****

>

> I suggest it is much more likely that the fiber/rear aperture coupling lens

> was at fault.  I must admit I don't see how with rear aperture illumination

> the same and same focal length for the 3 lasers  how the incident rays could

> have different angles ...

>

> Cheers

>

>

>

>  *****

>> To join, leave or search the confocal microscopy listserv, go to:

>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

>> *****

>>

>> Hi Michael,

>>

>> There are a few factors that need to be compromised in TIRF.   1) The TIRF

>> penetration depth is dependent upon the wavelength of the incident

>> illumination, the angle of incidence, and the refractive indices of the

>> media at the interface. 2) An apochromatic lens brings lasers of three

>> wavelengths on to the same focal plane, but may not to the same incident

>> angle.  Therefore, the incident angles for the three lasers should be able

>> to be adjusted independently to get optimal TIRF images.

>>

>> You may want to take your sample off first, to see how different the

>> angles

>> of the three lasers projecting to the wall, to get an idea of the incident

>> angles of your lasers coming out of the objective.

>>

>> Regards

>>

>> Gary G Li, PhD

>>

>>

>> On Sun, Jan 16, 2011 at 1:02 PM, Cammer, Michael <

>> [hidden email]

>>

>>> wrote:

>>>

>>

>>  *****

>>> To join, leave or search the confocal microscopy listserv, go to:

>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

>>> *****

>>>

>>> I originally posted the message below to the microscopy listserv but

>>> reviewing the discussion here on chromatic aberration, I thought it would

>>> be

>>> apropos here too even though it is not really regarding confocal.

>>>

>>> A few years ago we were having problems with the first commercial Olympus

>>> TIRF system because we could not get consistent evanescent waves with the

>>> one angle adjustment with the laser lines we had from 405 to 568 nm that

>>> were delivered via a single fiber (it was worse when we later added a 633

>>> nm

>>> laser).  I suggested we pump each laser in through a separate path that

>>> could be angled independently.  We didn't build it, but I think Olympus

>>> now

>>> sells a TIRF system that does this.

>>>

>>> Another issue is that when I first heard about TIRF maybe 15 years ago,

>>> it

>>> was introduced as a ring illumination at the outer edge of the back

>>> aperture, not as a single point or crescent at the periphery on only one

>>> side.  A ring, or at least a series of points around the periphery, seems

>>> like a better way to provide a uniform field due to aberrations from

>>> coherent light in the imperfect optics.  Any thought on this?

>>>

>>> Sincerely,

>>>

>>> Michael

>>>

>>> -----------------------ORIGINAL MESSAGE-------------------------------

>>> We have the Nikon TIRF system and have three laser lines

>>> going into the TIRF arm via a single fiber.  When we project through

>>> the 100X objective through the sample onto the wall we see that the

>>> lines go through the sample at different angles.  (You can see a

>>> picture of the projection at approx 45 degrees at

>>> http://www.flickr.com/photos/mcammer/5359189090/ .)  It is also

>>> noticeable in the TIRF images that the field depth is different for

>>> each wavelength.  Is this unavoidable due to the different

>>> wavelengths or is it possible to align the optics better so these

>>> spots would be more coincident?

>>>

>>>

>>>

>>> _________________________________________

>>> Michael Cammer, Assistant Research Scientist

>>> Skirball Institute of Biomolecular Medicine

>>> Lab: (212) 263-3208  Cell: (914) 309-3270

>>>

>>>

>>> ------------------------------------------------------------

>>> This email message, including any attachments, is for the sole use of the

>>> intended recipient(s) and may contain information that is proprietary,

>>> confidential, and exempt from disclosure under applicable law. Any

>>> unauthorized review, use, disclosure, or distribution is prohibited. If

>>> you

>>> have received this email in error please notify the sender by return

>>> email

>>> and delete the original message. Please note, the recipient should check

>>> this email and any attachments for the presence of viruses. The

>>> organization

>>> accepts no liability for any damage caused by any virus transmitted by

>>> this

>>> email.

>>> =================================

>>>

>>>


 
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Objective lens chromatic aberration

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Tube lens, or scan lens?

Craig


On Mon, Jan 17, 2011 at 3:06 PM, Andreas Bruckbauer <[hidden email]>wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> I guess the initial question about chromatic aberration and the new one
> about TIRF got muddeled up. I guess chromatically different NA in TIRF could
> arise when part of the correction is build in the tube lens? The
> illumination light forTIRF "through the objective" does not pass the tube
> lens and is thus is not well corrected. Furthermore slight imperfections in
> the dichroic which relfects the laser light can lead to distortions and
> these can be wavelength dependent.
>
>
> best wishes
>
> Andreas
>
>
>
> -----Original Message-----
> From: Craig Brideau <[hidden email]>
> To: [hidden email]
> Sent: Mon, 17 Jan 2011 19:01
> Subject: Re: Objective lens chromatic aberration
>
>
> *****
>
> To join, leave or search the confocal microscopy listserv, go to:
>
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>
> *****
>
>
>
> This would also lead to chromatically dependent NA, which would be a big
>
> no-no for a corrected objective.
>
>
>
> Craig
>
>
>
>
>
> On Sun, Jan 16, 2011 at 2:16 PM, Mark Cannell <[hidden email]
> >wrote:
>
>
>
> > *****
>
> > To join, leave or search the confocal microscopy listserv, go to:
>
> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>
> > *****
>
> >
>
> > I suggest it is much more likely that the fiber/rear aperture coupling
> lens
>
> > was at fault.  I must admit I don't see how with rear aperture
> illumination
>
> > the same and same focal length for the 3 lasers  how the incident rays
> could
>
> > have different angles ...
>
> >
>
> > Cheers
>
> >
>
> >
>
> >
>
> >  *****
>
> >> To join, leave or search the confocal microscopy listserv, go to:
>
> >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>
> >> *****
>
> >>
>
> >> Hi Michael,
>
> >>
>
> >> There are a few factors that need to be compromised in TIRF.   1) The
> TIRF
>
> >> penetration depth is dependent upon the wavelength of the incident
>
> >> illumination, the angle of incidence, and the refractive indices of the
>
> >> media at the interface. 2) An apochromatic lens brings lasers of three
>
> >> wavelengths on to the same focal plane, but may not to the same incident
>
> >> angle.  Therefore, the incident angles for the three lasers should be
> able
>
> >> to be adjusted independently to get optimal TIRF images.
>
> >>
>
> >> You may want to take your sample off first, to see how different the
>
> >> angles
>
> >> of the three lasers projecting to the wall, to get an idea of the
> incident
>
> >> angles of your lasers coming out of the objective.
>
> >>
>
> >> Regards
>
> >>
>
> >> Gary G Li, PhD
>
> >>
>
> >>
>
> >> On Sun, Jan 16, 2011 at 1:02 PM, Cammer, Michael <
>
> >> [hidden email]
>
> >>
>
> >>> wrote:
>
> >>>
>
> >>
>
> >>  *****
>
> >>> To join, leave or search the confocal microscopy listserv, go to:
>
> >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>
> >>> *****
>
> >>>
>
> >>> I originally posted the message below to the microscopy listserv but
>
> >>> reviewing the discussion here on chromatic aberration, I thought it
> would
>
> >>> be
>
> >>> apropos here too even though it is not really regarding confocal.
>
> >>>
>
> >>> A few years ago we were having problems with the first commercial
> Olympus
>
> >>> TIRF system because we could not get consistent evanescent waves with
> the
>
> >>> one angle adjustment with the laser lines we had from 405 to 568 nm
> that
>
> >>> were delivered via a single fiber (it was worse when we later added a
> 633
>
> >>> nm
>
> >>> laser).  I suggested we pump each laser in through a separate path that
>
> >>> could be angled independently.  We didn't build it, but I think Olympus
>
> >>> now
>
> >>> sells a TIRF system that does this.
>
> >>>
>
> >>> Another issue is that when I first heard about TIRF maybe 15 years ago,
>
> >>> it
>
> >>> was introduced as a ring illumination at the outer edge of the back
>
> >>> aperture, not as a single point or crescent at the periphery on only
> one
>
> >>> side.  A ring, or at least a series of points around the periphery,
> seems
>
> >>> like a better way to provide a uniform field due to aberrations from
>
> >>> coherent light in the imperfect optics.  Any thought on this?
>
> >>>
>
> >>> Sincerely,
>
> >>>
>
> >>> Michael
>
> >>>
>
> >>> -----------------------ORIGINAL MESSAGE-------------------------------
>
> >>> We have the Nikon TIRF system and have three laser lines
>
> >>> going into the TIRF arm via a single fiber.  When we project through
>
> >>> the 100X objective through the sample onto the wall we see that the
>
> >>> lines go through the sample at different angles.  (You can see a
>
> >>> picture of the projection at approx 45 degrees at
>
> >>> http://www.flickr.com/photos/mcammer/5359189090/ .)  It is also
>
> >>> noticeable in the TIRF images that the field depth is different for
>
> >>> each wavelength.  Is this unavoidable due to the different
>
> >>> wavelengths or is it possible to align the optics better so these
>
> >>> spots would be more coincident?
>
> >>>
>
> >>>
>
> >>>
>
> >>> _________________________________________
>
> >>> Michael Cammer, Assistant Research Scientist
>
> >>> Skirball Institute of Biomolecular Medicine
>
> >>> Lab: (212) 263-3208  Cell: (914) 309-3270
>
> >>>
>
> >>>
>
> >>> ------------------------------------------------------------
>
> >>> This email message, including any attachments, is for the sole use of
> the
>
> >>> intended recipient(s) and may contain information that is proprietary,
>
> >>> confidential, and exempt from disclosure under applicable law. Any
>
> >>> unauthorized review, use, disclosure, or distribution is prohibited. If
>
> >>> you
>
> >>> have received this email in error please notify the sender by return
>
> >>> email
>
> >>> and delete the original message. Please note, the recipient should
> check
>
> >>> this email and any attachments for the presence of viruses. The
>
> >>> organization
>
> >>> accepts no liability for any damage caused by any virus transmitted by
>
> >>> this
>
> >>> email.
>
> >>> =================================
>
> >>>
>
> >>>
>
>
>
>
Guy Cox-2 Guy Cox-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Objective lens chromatic aberration & TIRF

In reply to this post by Andreas Bruckbauer
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

What people seem to be missing in this thread is that glass and water
are both dispersive media so the critical angle will be different for
different wavelengths.  You cannot expect to get optimal TIRF with two
colours coming in at the same angle.  

 

 
Guy

 

Optical Imaging Techniques in Cell Biology

by Guy Cox    CRC Press / Taylor & Francis

     http://www.guycox.com/optical.htm
<http://www.guycox.com/optical.htm>

______________________________________________

Associate Professor Guy Cox, MA, DPhil(Oxon)

Australian Centre for Microscopy & Microanalysis,

Madsen Building F09, University of Sydney, NSW 2006

 

Phone +61 2 9351 3176     Fax +61 2 9351 7682

             Mobile 0413 281 861

______________________________________________

      http://www.guycox.net <http://www.guycox.net>

 

 

From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of Andreas Bruckbauer
Sent: Tuesday, 18 January 2011 9:06 AM
To: [hidden email]
Subject: Re: Objective lens chromatic aberration

 

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

I guess the initial question about chromatic aberration and the new one
about TIRF got muddeled up. I guess chromatically different NA in TIRF
could arise when part of the correction is build in the tube lens? The
illumination light forTIRF "through the objective" does not pass the
tube lens and is thus is not well corrected. Furthermore slight
imperfections in the dichroic which relfects the laser light can lead to
distortions and these can be wavelength dependent.  


best wishes

Andreas



-----Original Message-----
From: Craig Brideau <[hidden email]>
To: [hidden email]
Sent: Mon, 17 Jan 2011 19:01
Subject: Re: Objective lens chromatic aberration


*****

To join, leave or search the confocal microscopy listserv, go to:

http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

*****



This would also lead to chromatically dependent NA, which would be a big

no-no for a corrected objective.



Craig





On Sun, Jan 16, 2011 at 2:16 PM, Mark Cannell
<[hidden email]>wrote:



> *****

> To join, leave or search the confocal microscopy listserv, go to:

> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

> *****

>

> I suggest it is much more likely that the fiber/rear aperture coupling
lens

> was at fault.  I must admit I don't see how with rear aperture
illumination

> the same and same focal length for the 3 lasers  how the incident rays
could

> have different angles ...

>

> Cheers

>

>

>

>  *****

>> To join, leave or search the confocal microscopy listserv, go to:

>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

>> *****

>>

>> Hi Michael,

>>

>> There are a few factors that need to be compromised in TIRF.   1) The
TIRF

>> penetration depth is dependent upon the wavelength of the incident

>> illumination, the angle of incidence, and the refractive indices of
the

>> media at the interface. 2) An apochromatic lens brings lasers of
three

>> wavelengths on to the same focal plane, but may not to the same
incident

>> angle.  Therefore, the incident angles for the three lasers should be
able

>> to be adjusted independently to get optimal TIRF images.

>>

>> You may want to take your sample off first, to see how different the

>> angles

>> of the three lasers projecting to the wall, to get an idea of the
incident

>> angles of your lasers coming out of the objective.

>>

>> Regards

>>

>> Gary G Li, PhD

>>

>>

>> On Sun, Jan 16, 2011 at 1:02 PM, Cammer, Michael <

>> [hidden email]

>>

>>> wrote:

>>>

>>

>>  *****

>>> To join, leave or search the confocal microscopy listserv, go to:

>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

>>> *****

>>>

>>> I originally posted the message below to the microscopy listserv but

>>> reviewing the discussion here on chromatic aberration, I thought it
would

>>> be

>>> apropos here too even though it is not really regarding confocal.

>>>

>>> A few years ago we were having problems with the first commercial
Olympus

>>> TIRF system because we could not get consistent evanescent waves
with the

>>> one angle adjustment with the laser lines we had from 405 to 568 nm
that

>>> were delivered via a single fiber (it was worse when we later added
a 633

>>> nm

>>> laser).  I suggested we pump each laser in through a separate path
that

>>> could be angled independently.  We didn't build it, but I think
Olympus

>>> now

>>> sells a TIRF system that does this.

>>>

>>> Another issue is that when I first heard about TIRF maybe 15 years
ago,

>>> it

>>> was introduced as a ring illumination at the outer edge of the back

>>> aperture, not as a single point or crescent at the periphery on only
one

>>> side.  A ring, or at least a series of points around the periphery,
seems

>>> like a better way to provide a uniform field due to aberrations from

>>> coherent light in the imperfect optics.  Any thought on this?

>>>

>>> Sincerely,

>>>

>>> Michael

>>>

>>> -----------------------ORIGINAL
MESSAGE-------------------------------

>>> We have the Nikon TIRF system and have three laser lines

>>> going into the TIRF arm via a single fiber.  When we project through

>>> the 100X objective through the sample onto the wall we see that the

>>> lines go through the sample at different angles.  (You can see a

>>> picture of the projection at approx 45 degrees at

>>> http://www.flickr.com/photos/mcammer/5359189090/ .)  It is also

>>> noticeable in the TIRF images that the field depth is different for

>>> each wavelength.  Is this unavoidable due to the different

>>> wavelengths or is it possible to align the optics better so these

>>> spots would be more coincident?

>>>

>>>

>>>

>>> _________________________________________

>>> Michael Cammer, Assistant Research Scientist

>>> Skirball Institute of Biomolecular Medicine

>>> Lab: (212) 263-3208  Cell: (914) 309-3270

>>>

>>>

>>> ------------------------------------------------------------

>>> This email message, including any attachments, is for the sole use
of the

>>> intended recipient(s) and may contain information that is
proprietary,

>>> confidential, and exempt from disclosure under applicable law. Any

>>> unauthorized review, use, disclosure, or distribution is prohibited.
If

>>> you

>>> have received this email in error please notify the sender by return

>>> email

>>> and delete the original message. Please note, the recipient should
check

>>> this email and any attachments for the presence of viruses. The

>>> organization

>>> accepts no liability for any damage caused by any virus transmitted
by

>>> this

>>> email.

>>> =================================

>>>

>>>




________________________________

No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1191 / Virus Database: 1435/3342 - Release Date: 12/27/10
Internal Virus Database is out of date.
kspencer007 kspencer007
Reply | Threaded
Open this post in threaded view
♦
♦
|

tokai hit chamber

In reply to this post by Gary G. Li-2
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi all;
        Is anyone using a Tokai Hit incubation chamber for long-term live-cell imaging? May I contact you offline about humidity issues?
        Thanks.
        Kathy

        Kathy Spencer
        The Scripps Research Institute
        La Jolla, CA 92037
Ippei Kotera Ippei Kotera
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: tokai hit chamber

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Kathy,

Tokai Hit System I had been using was equipped with a doughnut-shaped
water chamber and heated lid along with CO2 and heat supplier. I have
done live imaging up to 5 hours with no humidity problem. What kind of
humidity issues do you have? The water chamber was simple enough that
nothing could go wrong, so it seemed.

Ippei

--
Ippei Kotera, Ph.D.

Tanz CRND
Tanz Neuroscience Building, Room 221
University of Toronto
6 Queen's Park Crescent West
Toronto, Ontario M5S 3H2
Phone: (416)978-2503
[hidden email]
Boswell, Carl A - (cboswell) Boswell, Carl A - (cboswell)
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: tokai hit chamber

In reply to this post by kspencer007
Hi Kathy,
If you're interested in offline conversations you might want to leave some  offline contact info.  I know of a unit being used here but I would need to defer to the person with experience with it.

Cheers,
C


Carl A. Boswell, Ph.D.
Molecular and Cellular Biology
Univ. of Arizona
520-954-7053
FAX 520-621-3709


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Kathryn Spencer
Sent: Thursday, February 10, 2011 3:14 PM
To: [hidden email]
Subject: tokai hit chamber

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi all;
        Is anyone using a Tokai Hit incubation chamber for long-term live-cell imaging? May I contact you offline about humidity issues?
        Thanks.
        Kathy

        Kathy Spencer
        The Scripps Research Institute
        La Jolla, CA 92037

kspencer007 kspencer007
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: tokai hit chamber

In reply to this post by Ippei Kotera
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Thanks Ippei;
        We are imaging 2-3 days, and the reservoir is drying out. Short term, it does seem OK. I'm just unsure about why it is drying out, if we are just imaging much longer than normal. If we put in wet Kimwipes, everything seems fine. Somehow, we need to add water during our imaging.
        Kathy


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ippei Kotera
Sent: Thursday, February 10, 2011 3:56 PM
To: [hidden email]
Subject: Re: tokai hit chamber

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Kathy,

Tokai Hit System I had been using was equipped with a doughnut-shaped
water chamber and heated lid along with CO2 and heat supplier. I have
done live imaging up to 5 hours with no humidity problem. What kind of
humidity issues do you have? The water chamber was simple enough that
nothing could go wrong, so it seemed.

Ippei

--
Ippei Kotera, Ph.D.

Tanz CRND
Tanz Neuroscience Building, Room 221
University of Toronto
6 Queen's Park Crescent West
Toronto, Ontario M5S 3H2
Phone: (416)978-2503
[hidden email]
Romin, Yevgeniy/Sloan Kettering Institute Romin, Yevgeniy/Sloan Kettering Institute
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: tokai hit chamber

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Kathy

We have several Tokai Hit icubators in our facility, they all dry out within 24 hours, the company seems to be aware of it and they suggest to put in more water every 12 hours, so that is what we do.  Wet Kimwipes also help.  Or wet cotton balls.

________________________________________
From: Confocal Microscopy List [[hidden email]] On Behalf Of Kathryn Spencer [[hidden email]]
Sent: Thursday, February 10, 2011 7:17 PM
To: [hidden email]
Subject: Re: tokai hit chamber

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Thanks Ippei;
        We are imaging 2-3 days, and the reservoir is drying out. Short term, it does seem OK. I'm just unsure about why it is drying out, if we are just imaging much longer than normal. If we put in wet Kimwipes, everything seems fine. Somehow, we need to add water during our imaging.
        Kathy


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ippei Kotera
Sent: Thursday, February 10, 2011 3:56 PM
To: [hidden email]
Subject: Re: tokai hit chamber

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Kathy,

Tokai Hit System I had been using was equipped with a doughnut-shaped
water chamber and heated lid along with CO2 and heat supplier. I have
done live imaging up to 5 hours with no humidity problem. What kind of
humidity issues do you have? The water chamber was simple enough that
nothing could go wrong, so it seemed.

Ippei

--
Ippei Kotera, Ph.D.

Tanz CRND
Tanz Neuroscience Building, Room 221
University of Toronto
6 Queen's Park Crescent West
Toronto, Ontario M5S 3H2
Phone: (416)978-2503
[hidden email]


 
     =====================================================================
     
     Please note that this e-mail and any files transmitted with it may be
     privileged, confidential, and protected from disclosure under
     applicable law. If the reader of this message is not the intended
     recipient, or an employee or agent responsible for delivering this
     message to the intended recipient, you are hereby notified that any
     reading, dissemination, distribution, copying, or other use of this
     communication or any of its attachments is strictly prohibited.  If
     you have received this communication in error, please notify the
     sender immediately by replying to this message and deleting this
     message, any attachments, and all copies and backups from your
     computer.
kspencer007 kspencer007
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: tokai hit chamber

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Great advice. Thank you.
        Kathy


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Yevgeniy Romin
Sent: Thursday, February 10, 2011 5:54 PM
To: [hidden email]
Subject: Re: tokai hit chamber

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Kathy

We have several Tokai Hit icubators in our facility, they all dry out within 24 hours, the company seems to be aware of it and they suggest to put in more water every 12 hours, so that is what we do.  Wet Kimwipes also help.  Or wet cotton balls.

________________________________________
From: Confocal Microscopy List [[hidden email]] On Behalf Of Kathryn Spencer [[hidden email]]
Sent: Thursday, February 10, 2011 7:17 PM
To: [hidden email]
Subject: Re: tokai hit chamber

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Thanks Ippei;
        We are imaging 2-3 days, and the reservoir is drying out. Short term, it does seem OK. I'm just unsure about why it is drying out, if we are just imaging much longer than normal. If we put in wet Kimwipes, everything seems fine. Somehow, we need to add water during our imaging.
        Kathy


-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ippei Kotera
Sent: Thursday, February 10, 2011 3:56 PM
To: [hidden email]
Subject: Re: tokai hit chamber

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Kathy,

Tokai Hit System I had been using was equipped with a doughnut-shaped
water chamber and heated lid along with CO2 and heat supplier. I have
done live imaging up to 5 hours with no humidity problem. What kind of
humidity issues do you have? The water chamber was simple enough that
nothing could go wrong, so it seemed.

Ippei

--
Ippei Kotera, Ph.D.

Tanz CRND
Tanz Neuroscience Building, Room 221
University of Toronto
6 Queen's Park Crescent West
Toronto, Ontario M5S 3H2
Phone: (416)978-2503
[hidden email]


 
     =====================================================================
     
     Please note that this e-mail and any files transmitted with it may be
     privileged, confidential, and protected from disclosure under
     applicable law. If the reader of this message is not the intended
     recipient, or an employee or agent responsible for delivering this
     message to the intended recipient, you are hereby notified that any
     reading, dissemination, distribution, copying, or other use of this
     communication or any of its attachments is strictly prohibited.  If
     you have received this communication in error, please notify the
     sender immediately by replying to this message and deleting this
     message, any attachments, and all copies and backups from your
     computer.
Knecht, David Knecht, David
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: tokai hit chamber

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

FYI- Our Pathology Devices stage incubator does not have this problem and works for days with little media loss.  Dave

On Feb 10, 2011, at 10:06 PM, Kathryn Spencer wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Great advice. Thank you.
> Kathy
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Yevgeniy Romin
> Sent: Thursday, February 10, 2011 5:54 PM
> To: [hidden email]
> Subject: Re: tokai hit chamber
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Kathy
>
> We have several Tokai Hit icubators in our facility, they all dry out within 24 hours, the company seems to be aware of it and they suggest to put in more water every 12 hours, so that is what we do.  Wet Kimwipes also help.  Or wet cotton balls.
>
> ________________________________________
> From: Confocal Microscopy List [[hidden email]] On Behalf Of Kathryn Spencer [[hidden email]]
> Sent: Thursday, February 10, 2011 7:17 PM
> To: [hidden email]
> Subject: Re: tokai hit chamber
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Thanks Ippei;
>        We are imaging 2-3 days, and the reservoir is drying out. Short term, it does seem OK. I'm just unsure about why it is drying out, if we are just imaging much longer than normal. If we put in wet Kimwipes, everything seems fine. Somehow, we need to add water during our imaging.
>        Kathy
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ippei Kotera
> Sent: Thursday, February 10, 2011 3:56 PM
> To: [hidden email]
> Subject: Re: tokai hit chamber
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Kathy,
>
> Tokai Hit System I had been using was equipped with a doughnut-shaped
> water chamber and heated lid along with CO2 and heat supplier. I have
> done live imaging up to 5 hours with no humidity problem. What kind of
> humidity issues do you have? The water chamber was simple enough that
> nothing could go wrong, so it seemed.
>
> Ippei
>
> --
> Ippei Kotera, Ph.D.
>
> Tanz CRND
> Tanz Neuroscience Building, Room 221
> University of Toronto
> 6 Queen's Park Crescent West
> Toronto, Ontario M5S 3H2
> Phone: (416)978-2503
> [hidden email]
>
>
>
>     =====================================================================
>
>     Please note that this e-mail and any files transmitted with it may be
>     privileged, confidential, and protected from disclosure under
>     applicable law. If the reader of this message is not the intended
>     recipient, or an employee or agent responsible for delivering this
>     message to the intended recipient, you are hereby notified that any
>     reading, dissemination, distribution, copying, or other use of this
>     communication or any of its attachments is strictly prohibited.  If
>     you have received this communication in error, please notify the
>     sender immediately by replying to this message and deleting this
>     message, any attachments, and all copies and backups from your
>     computer.

Dr. David Knecht    
Department of Molecular and Cell Biology
Co-head Flow Cytometry and Confocal Microscopy Facility
U-3125
91 N. Eagleville Rd.
University of Connecticut
Storrs, CT 06269
860-486-2200
860-486-4331 (fax)
Tim Feinstein-2 Tim Feinstein-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: tokai hit chamber

In reply to this post by kspencer007
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Kathy,

Hi Kathy,

If you have a 5% CO2 gas flow into the chamber, have you tried bubbling it through a small reservoir of distilled water?  Tank gas is quite dry.  A while back I used that trick in a jerryrigged unit for 24-h. imaging.  We have a Tokai-Hit now but have not yet needed it for very long-term imaging.  

cheers,


Tim Feinstein


On Feb 10, 2011, at 7:17 PM, Kathryn Spencer wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Thanks Ippei;
> We are imaging 2-3 days, and the reservoir is drying out. Short term, it does seem OK. I'm just unsure about why it is drying out, if we are just imaging much longer than normal. If we put in wet Kimwipes, everything seems fine. Somehow, we need to add water during our imaging.
> Kathy
>
>
> -----Original Message-----
> From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Ippei Kotera
> Sent: Thursday, February 10, 2011 3:56 PM
> To: [hidden email]
> Subject: Re: tokai hit chamber
>
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Kathy,
>
> Tokai Hit System I had been using was equipped with a doughnut-shaped
> water chamber and heated lid along with CO2 and heat supplier. I have
> done live imaging up to 5 hours with no humidity problem. What kind of