						Kyle Michael Douglass | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi everyone, A couple years ago there was a post on this forum about significant astigmatism appearing in the reflected path of a custom-built, two camera TIRF setup. (http://confocal-microscopy-list.588098.n2.nabble.com/Reflected-image-gt-astigmatism-td7584402.html) The original poster had placed a dichroic before a pair of matched tube lenses to split the two color channels onto two separate cameras. The general consensus was that you could "buy your way out of the problem" by buying a thick, ultraflat dichroic. I am trying a similar approach on a setup in our lab but, due to space constraints that are not easily overcome, have tried first placing the channel-splitting dichroic in the image space after the final tube lens instead of in the infinity space between the objective and tube lens. Even with a 3 mm thick ultraflat dichroic, I see moderately bad astigmatism in the reflected channel. I have ruled out other possible sources of the astigmatism, like an additional dichroic for the excitation light and filters that already lie in the infinity space between the objective and tube lens. My question is: is it even worth trying to eliminate the astigmatism in the reflected path of a split-channel setup if the dichroic does not lie in the infinity space? Or is it pretty much always a bad idea to put the splitter dichroic in the image space? My current suspicion is that the tolerances in the alignment have to be very, very tight to avoid astigmatism with the splitter in the image space. Thanks for your feedback. Kyle -- Kyle M. Douglass, PhD Post-doctoral researcher The Laboratory of Experimental Biophysics EPFL, Lausanne, Switzerland http://kmdouglass.github.io http://leb.epfl.ch  | 
			
 
	
	 | 
						Zdenek Svindrych-2 | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Kyle, if your transmitted image is free of astigmatism, then the dichroic must be responsible for the astigmatism in the reflected path (assuming there are no other components between the dichoic and the cameras). Try a flat mirror (I mean really flat, 6mm substrate, lambda/10 specs) instead of the dichroic and see if the astigmatism is gone. The new 3mm dichroics have similar specs (lambda/4 to lambda/10), so if it is mounted properly (it's not about exact angles, but you must not introduce any stress that may lead to bending), it should work OK. Try another dichroic, maybe one factory-mounted in a cube. Also, the closer the dichroic to the camera, the smaller the effects of curvature. I don't think infinty space is less 'sensitive' to astigmatism. Quite the opposite - least sensitive should be a position conjugate to image plane (but then, every piece of dust will be in focus on the camera...). Best, zdenek Zdenek Svindrych, Ph.D. W.M. Keck Center for Cellular Imaging (PLSB 003) Department of Biology,University of Virginia 409 McCormick Rd, Charlottesville, VA-22904 http://www.kcci.virginia.edu/ tel: 434-982-4869 Annual FRET Workshop (March 6-10, 2017): http://kcci.virginia.edu/workshop- 2017 ---------- Původní zpráva ---------- Od: Kyle Douglass <[hidden email]> Komu: [hidden email] Datum: 13. 2. 2017 9:29:31 Předmět: Reflected image -> astigmatism "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi everyone, A couple years ago there was a post on this forum about significant astigmatism appearing in the reflected path of a custom-built, two camera TIRF setup. (http://confocal-microscopy-list.588098.n2.nabble.com/Reflected-image-gt- astigmatism-td7584402.html) The original poster had placed a dichroic before a pair of matched tube lenses to split the two color channels onto two separate cameras. The general consensus was that you could "buy your way out of the problem" by buying a thick, ultraflat dichroic. I am trying a similar approach on a setup in our lab but, due to space constraints that are not easily overcome, have tried first placing the channel-splitting dichroic in the image space after the final tube lens instead of in the infinity space between the objective and tube lens. Even with a 3 mm thick ultraflat dichroic, I see moderately bad astigmatism in the reflected channel. I have ruled out other possible sources of the astigmatism, like an additional dichroic for the excitation light and filters that already lie in the infinity space between the objective and tube lens. My question is: is it even worth trying to eliminate the astigmatism in the reflected path of a split-channel setup if the dichroic does not lie in the infinity space? Or is it pretty much always a bad idea to put the splitter dichroic in the image space? My current suspicion is that the tolerances in the alignment have to be very, very tight to avoid astigmatism with the splitter in the image space. Thanks for your feedback. Kyle -- Kyle M. Douglass, PhD Post-doctoral researcher The Laboratory of Experimental Biophysics EPFL, Lausanne, Switzerland http://kmdouglass.github.io http://leb.epfl.ch "  | 
			
 
	
	 | 
						Jeffrey Carmichael | 
| 
					
		 | 
				
					
	 
				In reply to this post by Kyle Michael Douglass
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** COMMERCIAL RESPONSE Hi Kyle, We've found that it's most often the mounting of the dichroic itself that causes the most distortion in an image reflected off of a dichroic. You could theoretically have an "ideal" dichroic with no curvature, but once it is affixed, it will be torqued out of flatness to some extent unless you can completely avoid any mounting stress. If held by means that apply pressure on top or bottom of the large surface, you will get various forms of non-spherical astigmatism. Even light pressure on the sides of the dichroic will have some effect. As Zdenek suggested, thicker dichroics are stiffer and will better resist deformation. If possible, try using something like silicone RTV applied very sparingly in small beads around the outside edges of the dichroic, but never under it. This serves to prevent the dichroic from moving around without applying stress. If the "holder" is removable, then you can swap out these mounted dichroics and not fiddle with the dichroic itself once affixed. Jeff *Jeff Carmichael* *Technical and Product Marketing Manager* *[hidden email] <[hidden email]>*Chroma Technology Corp. *an employee owned company* *10 Imtec Lane* *Bellows Falls, VT 05301* *802-428-2528 Office* *802-428-2528 Fax**800-824-7662 Toll Free* On Mon, Feb 13, 2017 at 9:27 AM, Kyle Douglass <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi everyone, > > A couple years ago there was a post on this forum about significant > astigmatism appearing in the reflected path of a custom-built, two camera > TIRF setup. (http://confocal-microscopy-list.588098.n2.nabble.com/Reflec > ted-image-gt-astigmatism-td7584402.html) The original poster had placed a > dichroic before a pair of matched tube lenses to split the two color > channels onto two separate cameras. The general consensus was that you > could "buy your way out of the problem" by buying a thick, ultraflat > dichroic. > > I am trying a similar approach on a setup in our lab but, due to space > constraints that are not easily overcome, have tried first placing the > channel-splitting dichroic in the image space after the final tube lens > instead of in the infinity space between the objective and tube lens. Even > with a 3 mm thick ultraflat dichroic, I see moderately bad astigmatism in > the reflected channel. I have ruled out other possible sources of the > astigmatism, like an additional dichroic for the excitation light and > filters that already lie in the infinity space between the objective and > tube lens. > > My question is: is it even worth trying to eliminate the astigmatism in > the reflected path of a split-channel setup if the dichroic does not lie in > the infinity space? Or is it pretty much always a bad idea to put the > splitter dichroic in the image space? > > My current suspicion is that the tolerances in the alignment have to be > very, very tight to avoid astigmatism with the splitter in the image space. > > Thanks for your feedback. > > Kyle > > -- > Kyle M. Douglass, PhD > Post-doctoral researcher > The Laboratory of Experimental Biophysics > EPFL, Lausanne, Switzerland > http://kmdouglass.github.io > http://leb.epfl.ch >  | 
			
 
	
	 | 
						Rusty Nicovich | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Kyle, The other poster was me. I ultimately fixed the issue to my satisfaction by carefully mounting the dichroic to the cube with double-sided tape rather than the provided clamp. That method was stress-free enough to not induce any undue curvature in the dichroic, at least qualitatively as no astigmatism immediately obvious in the reflected image. This was even with a relatively thin (1.1 mm) substrate for the mirror. This solution is similar to what Jeff suggested using silicone adhesive instead of tape. Another fun part of that system is that the camera fans would induce quite a large vibration and only along one axis. The vibration period was 10-20 ms, which meant any longer exposures would yield a PSF elongated on that axis. Astigmatism would vary along the focal axis and this didn't, but it still gave an asymmetric PSF at the focal plane. Thanks, Rusty On Mon, Feb 13, 2017 at 7:49 AM, Jeffrey Carmichael <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > COMMERCIAL RESPONSE > > Hi Kyle, > > We've found that it's most often the mounting of the dichroic itself that > causes the most distortion in an image reflected off of a dichroic. You > could theoretically have an "ideal" dichroic with no curvature, but once it > is affixed, it will be torqued out of flatness to some extent unless you > can completely avoid any mounting stress. > > If held by means that apply pressure on top or bottom of the large surface, > you will get various forms of non-spherical astigmatism. Even light > pressure on the sides of the dichroic will have some effect. > > As Zdenek suggested, thicker dichroics are stiffer and will better resist > deformation. > > If possible, try using something like silicone RTV applied very sparingly > in small beads around the outside edges of the dichroic, but never under > it. This serves to prevent the dichroic from moving around without > applying stress. If the "holder" is removable, then you can swap out these > mounted dichroics and not fiddle with the dichroic itself once affixed. > > Jeff > > > *Jeff Carmichael* > > *Technical and Product Marketing Manager* > > *[hidden email] <[hidden email]>*Chroma Technology Corp. > > *an employee owned company* > *10 Imtec Lane* > *Bellows Falls, VT 05301* > *802-428-2528 Office* > *802-428-2528 Fax**800-824-7662 Toll Free* > > On Mon, Feb 13, 2017 at 9:27 AM, Kyle Douglass <[hidden email]> > wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Hi everyone, > > > > A couple years ago there was a post on this forum about significant > > astigmatism appearing in the reflected path of a custom-built, two camera > > TIRF setup. (http://confocal-microscopy-list.588098.n2.nabble.com/Reflec > > ted-image-gt-astigmatism-td7584402.html) The original poster had placed > a > > dichroic before a pair of matched tube lenses to split the two color > > channels onto two separate cameras. The general consensus was that you > > could "buy your way out of the problem" by buying a thick, ultraflat > > dichroic. > > > > I am trying a similar approach on a setup in our lab but, due to space > > constraints that are not easily overcome, have tried first placing the > > channel-splitting dichroic in the image space after the final tube lens > > instead of in the infinity space between the objective and tube lens. > Even > > with a 3 mm thick ultraflat dichroic, I see moderately bad astigmatism in > > the reflected channel. I have ruled out other possible sources of the > > astigmatism, like an additional dichroic for the excitation light and > > filters that already lie in the infinity space between the objective and > > tube lens. > > > > My question is: is it even worth trying to eliminate the astigmatism in > > the reflected path of a split-channel setup if the dichroic does not lie > in > > the infinity space? Or is it pretty much always a bad idea to put the > > splitter dichroic in the image space? > > > > My current suspicion is that the tolerances in the alignment have to be > > very, very tight to avoid astigmatism with the splitter in the image > space. > > > > Thanks for your feedback. > > > > Kyle > > > > -- > > Kyle M. Douglass, PhD > > Post-doctoral researcher > > The Laboratory of Experimental Biophysics > > EPFL, Lausanne, Switzerland > > http://kmdouglass.github.io > > http://leb.epfl.ch > > >  | 
			
 
	
	 | 
						Andrew York | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I'm worried that your transmission path will also have astigmatism. Even a perfectly flat piece of glass yields astigmatism in transmission if it's tilted in a non-infinity space, right? (Warning: lousy advice below) I've encountered this myself, and "fixed" it with a second piece of glass, the same thickness and material as the first, tilted equally about an orthogonal axis, to introduce "equal and opposite" astigmatism. This was a hack, but it was good enough for my needs at the time. Depending on your needs, perhaps you could use tilted flat glass as a tunable astigmatism compensation. On Mon, Feb 13, 2017 at 8:45 AM, Rusty Nicovich <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Kyle, > > The other poster was me. I ultimately fixed the issue to my satisfaction > by carefully mounting the dichroic to the cube with double-sided tape > rather than the provided clamp. That method was stress-free enough to not > induce any undue curvature in the dichroic, at least qualitatively as no > astigmatism immediately obvious in the reflected image. This was even with > a relatively thin (1.1 mm) substrate for the mirror. This solution is > similar to what Jeff suggested using silicone adhesive instead of tape. > > Another fun part of that system is that the camera fans would induce quite > a large vibration and only along one axis. The vibration period was 10-20 > ms, which meant any longer exposures would yield a PSF elongated on that > axis. Astigmatism would vary along the focal axis and this didn't, but it > still gave an asymmetric PSF at the focal plane. > > Thanks, > Rusty > > On Mon, Feb 13, 2017 at 7:49 AM, Jeffrey Carmichael < > [hidden email]> > wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > COMMERCIAL RESPONSE > > > > Hi Kyle, > > > > We've found that it's most often the mounting of the dichroic itself that > > causes the most distortion in an image reflected off of a dichroic. You > > could theoretically have an "ideal" dichroic with no curvature, but once > it > > is affixed, it will be torqued out of flatness to some extent unless you > > can completely avoid any mounting stress. > > > > If held by means that apply pressure on top or bottom of the large > surface, > > you will get various forms of non-spherical astigmatism. Even light > > pressure on the sides of the dichroic will have some effect. > > > > As Zdenek suggested, thicker dichroics are stiffer and will better resist > > deformation. > > > > If possible, try using something like silicone RTV applied very sparingly > > in small beads around the outside edges of the dichroic, but never under > > it. This serves to prevent the dichroic from moving around without > > applying stress. If the "holder" is removable, then you can swap out > these > > mounted dichroics and not fiddle with the dichroic itself once affixed. > > > > Jeff > > > > > > *Jeff Carmichael* > > > > *Technical and Product Marketing Manager* > > > > *[hidden email] <[hidden email]>*Chroma Technology Corp. > > > > *an employee owned company* > > *10 Imtec Lane* > > *Bellows Falls, VT 05301* > > *802-428-2528 Office* > > *802-428-2528 Fax**800-824-7662 Toll Free* > > > > On Mon, Feb 13, 2017 at 9:27 AM, Kyle Douglass <[hidden email]> > > wrote: > > > > > ***** > > > To join, leave or search the confocal microscopy listserv, go to: > > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > > Post images on http://www.imgur.com and include the link in your > > posting. > > > ***** > > > > > > Hi everyone, > > > > > > A couple years ago there was a post on this forum about significant > > > astigmatism appearing in the reflected path of a custom-built, two > camera > > > TIRF setup. (http://confocal-microscopy-list.588098.n2.nabble.com/ > Reflec > > > ted-image-gt-astigmatism-td7584402.html) The original poster had > placed > > a > > > dichroic before a pair of matched tube lenses to split the two color > > > channels onto two separate cameras. The general consensus was that you > > > could "buy your way out of the problem" by buying a thick, ultraflat > > > dichroic. > > > > > > I am trying a similar approach on a setup in our lab but, due to space > > > constraints that are not easily overcome, have tried first placing the > > > channel-splitting dichroic in the image space after the final tube lens > > > instead of in the infinity space between the objective and tube lens. > > Even > > > with a 3 mm thick ultraflat dichroic, I see moderately bad astigmatism > in > > > the reflected channel. I have ruled out other possible sources of the > > > astigmatism, like an additional dichroic for the excitation light and > > > filters that already lie in the infinity space between the objective > and > > > tube lens. > > > > > > My question is: is it even worth trying to eliminate the astigmatism in > > > the reflected path of a split-channel setup if the dichroic does not > lie > > in > > > the infinity space? Or is it pretty much always a bad idea to put the > > > splitter dichroic in the image space? > > > > > > My current suspicion is that the tolerances in the alignment have to be > > > very, very tight to avoid astigmatism with the splitter in the image > > space. > > > > > > Thanks for your feedback. > > > > > > Kyle > > > > > > -- > > > Kyle M. Douglass, PhD > > > Post-doctoral researcher > > > The Laboratory of Experimental Biophysics > > > EPFL, Lausanne, Switzerland > > > http://kmdouglass.github.io > > > http://leb.epfl.ch > > > > > >  | 
			
 
	
	 | 
						Kyle Michael Douglass | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi everyone, Thanks a lot for your replies. To summarize your responses as I understand them so far: 1) Mounting of the dichroic is an absolutely critical step and first thing to check when troubleshooting astigmatism. I was taking care to tighten the set screws in the mount only to the extent required to prevent the dichroic from falling out. However, it seems that even this is suspect and can warp the dichroic surface enough to induce astigmatism upon reflection. 2) Using double-sided tape or a silicone adhesive applied to the outside edges of the dichroic is a better way to mount it; avoid applying any pressure to the dichroics large front and back surfaces. 3) Using a flat mirror in place of the dichroic is one way to check whether it's really the dichroic causing the astigmatism or something else. 4) Based on differing responses, it's still not clear to me whether a splitter dichroic in the image space (as opposed to the infinity space) is a workable solution or something to be avoided at all costs. I can perhaps partially correct for the astigmatism with a flat compensator plate. Now that I think about it, I had already known that a tilted coverslip can induce coma, especially in water-immersion objectives (http://onlinelibrary.wiley.com/doi/10.1111/j.0022-2720.2004.01383.x/abstract <http://onlinelibrary.wiley.com/doi/10.1111/j.0022-2720.2004.01383.x/pdf>). In general, coma (and astigmatism too, I think) originates from any element where rays from common object points but traveling at different azimuthal angles see different path lengths. I don't see why this logic couldn't extend to a tilted dichroic in the image space as well. Best regards, Kyle On 02/13/2017 10:34 PM, Andrew York wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I'm worried that your transmission path will also have astigmatism. Even a > perfectly flat piece of glass yields astigmatism in transmission if it's > tilted in a non-infinity space, right? > > (Warning: lousy advice below) > I've encountered this myself, and "fixed" it with a second piece of glass, > the same thickness and material as the first, tilted equally about an > orthogonal axis, to introduce "equal and opposite" astigmatism. This was a > hack, but it was good enough for my needs at the time. Depending on your > needs, perhaps you could use tilted flat glass as a tunable astigmatism > compensation. > > On Mon, Feb 13, 2017 at 8:45 AM, Rusty Nicovich <[hidden email]> > wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Kyle, >> >> The other poster was me. I ultimately fixed the issue to my satisfaction >> by carefully mounting the dichroic to the cube with double-sided tape >> rather than the provided clamp. That method was stress-free enough to not >> induce any undue curvature in the dichroic, at least qualitatively as no >> astigmatism immediately obvious in the reflected image. This was even with >> a relatively thin (1.1 mm) substrate for the mirror. This solution is >> similar to what Jeff suggested using silicone adhesive instead of tape. >> >> Another fun part of that system is that the camera fans would induce quite >> a large vibration and only along one axis. The vibration period was 10-20 >> ms, which meant any longer exposures would yield a PSF elongated on that >> axis. Astigmatism would vary along the focal axis and this didn't, but it >> still gave an asymmetric PSF at the focal plane. >> >> Thanks, >> Rusty >> >> On Mon, Feb 13, 2017 at 7:49 AM, Jeffrey Carmichael < >> [hidden email]> >> wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> Post images on http://www.imgur.com and include the link in your >> posting. >>> ***** >>> >>> COMMERCIAL RESPONSE >>> >>> Hi Kyle, >>> >>> We've found that it's most often the mounting of the dichroic itself that >>> causes the most distortion in an image reflected off of a dichroic. You >>> could theoretically have an "ideal" dichroic with no curvature, but once >> it >>> is affixed, it will be torqued out of flatness to some extent unless you >>> can completely avoid any mounting stress. >>> >>> If held by means that apply pressure on top or bottom of the large >> surface, >>> you will get various forms of non-spherical astigmatism. Even light >>> pressure on the sides of the dichroic will have some effect. >>> >>> As Zdenek suggested, thicker dichroics are stiffer and will better resist >>> deformation. >>> >>> If possible, try using something like silicone RTV applied very sparingly >>> in small beads around the outside edges of the dichroic, but never under >>> it. This serves to prevent the dichroic from moving around without >>> applying stress. If the "holder" is removable, then you can swap out >> these >>> mounted dichroics and not fiddle with the dichroic itself once affixed. >>> >>> Jeff >>> >>> >>> *Jeff Carmichael* >>> >>> *Technical and Product Marketing Manager* >>> >>> *[hidden email] <[hidden email]>*Chroma Technology Corp. >>> >>> *an employee owned company* >>> *10 Imtec Lane* >>> *Bellows Falls, VT 05301* >>> *802-428-2528 Office* >>> *802-428-2528 Fax**800-824-7662 Toll Free* >>> >>> On Mon, Feb 13, 2017 at 9:27 AM, Kyle Douglass <[hidden email]> >>> wrote: >>> >>>> ***** >>>> To join, leave or search the confocal microscopy listserv, go to: >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>> Post images on http://www.imgur.com and include the link in your >>> posting. >>>> ***** >>>> >>>> Hi everyone, >>>> >>>> A couple years ago there was a post on this forum about significant >>>> astigmatism appearing in the reflected path of a custom-built, two >> camera >>>> TIRF setup. (http://confocal-microscopy-list.588098.n2.nabble.com/ >> Reflec >>>> ted-image-gt-astigmatism-td7584402.html) The original poster had >> placed >>> a >>>> dichroic before a pair of matched tube lenses to split the two color >>>> channels onto two separate cameras. The general consensus was that you >>>> could "buy your way out of the problem" by buying a thick, ultraflat >>>> dichroic. >>>> >>>> I am trying a similar approach on a setup in our lab but, due to space >>>> constraints that are not easily overcome, have tried first placing the >>>> channel-splitting dichroic in the image space after the final tube lens >>>> instead of in the infinity space between the objective and tube lens. >>> Even >>>> with a 3 mm thick ultraflat dichroic, I see moderately bad astigmatism >> in >>>> the reflected channel. I have ruled out other possible sources of the >>>> astigmatism, like an additional dichroic for the excitation light and >>>> filters that already lie in the infinity space between the objective >> and >>>> tube lens. >>>> >>>> My question is: is it even worth trying to eliminate the astigmatism in >>>> the reflected path of a split-channel setup if the dichroic does not >> lie >>> in >>>> the infinity space? Or is it pretty much always a bad idea to put the >>>> splitter dichroic in the image space? >>>> >>>> My current suspicion is that the tolerances in the alignment have to be >>>> very, very tight to avoid astigmatism with the splitter in the image >>> space. >>>> Thanks for your feedback. >>>> >>>> Kyle >>>> >>>> -- >>>> Kyle M. Douglass, PhD >>>> Post-doctoral researcher >>>> The Laboratory of Experimental Biophysics >>>> EPFL, Lausanne, Switzerland >>>> http://kmdouglass.github.io >>>> http://leb.epfl.ch >>>> -- Kyle M. Douglass, PhD Post-doctoral researcher The Laboratory of Experimental Biophysics EPFL, Lausanne, Switzerland http://kmdouglass.github.io http://leb.epfl.ch  | 
			
 
	
	 | 
						Zdenek Svindrych-2 | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Kyle, First, reflection off of a flat mirror does not introduce any aberrations. That's clear. So as long as it's the first surface that is reflective, the reflected path should be perfect. Second, transmission through a flat glass slab introduces spherical aberration (and other aberrations if it's tilted). But the camera space numerical aperture is so low that you can ignore them. So, the main reason against putting filters and dichroics into the camera image space (as I see it) is image shift and defocus as you change the thickness and angle of the filters - this only applies to filter wheels, not to rigidly fixed setup. Secondary reasons are blemishes (dirt on the filters may be visible on the camera with very low NA illumination, especially when observing specimens with transmitted light) and filter size (usually the camera chip is larger than the back focal aperture of the objective lens, so in principle you could use smaller filter there - not a concern in modern microscopes where all filters are 1 inch dia...). Best, zdenek -- Zdenek Svindrych, Ph.D. W.M. Keck Center for Cellular Imaging (PLSB 003) Department of Biology,University of Virginia 409 McCormick Rd, Charlottesville, VA-22904 http://www.kcci.virginia.edu/ tel: 434-982-4869 Annual FRET Workshop (March 6-10, 2017): http://kcci.virginia.edu/workshop- 2017 ---------- Původní zpráva ---------- Od: Kyle Douglass <[hidden email]> Komu: [hidden email] Datum: 14. 2. 2017 3:21:53 Předmět: Re: Reflected image -> astigmatism "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi everyone, Thanks a lot for your replies. To summarize your responses as I understand them so far: 1) Mounting of the dichroic is an absolutely critical step and first thing to check when troubleshooting astigmatism. I was taking care to tighten the set screws in the mount only to the extent required to prevent the dichroic from falling out. However, it seems that even this is suspect and can warp the dichroic surface enough to induce astigmatism upon reflection. 2) Using double-sided tape or a silicone adhesive applied to the outside edges of the dichroic is a better way to mount it; avoid applying any pressure to the dichroics large front and back surfaces. 3) Using a flat mirror in place of the dichroic is one way to check whether it's really the dichroic causing the astigmatism or something else. 4) Based on differing responses, it's still not clear to me whether a splitter dichroic in the image space (as opposed to the infinity space) is a workable solution or something to be avoided at all costs. I can perhaps partially correct for the astigmatism with a flat compensator plate. Now that I think about it, I had already known that a tilted coverslip can induce coma, especially in water-immersion objectives (http://onlinelibrary.wiley.com/doi/10.1111/j.0022-2720.2004.01383.x/ abstract <http://onlinelibrary.wiley.com/doi/10.1111/j.0022-2720.2004.01383.x/pdf>). In general, coma (and astigmatism too, I think) originates from any element where rays from common object points but traveling at different azimuthal angles see different path lengths. I don't see why this logic couldn't extend to a tilted dichroic in the image space as well. Best regards, Kyle On 02/13/2017 10:34 PM, Andrew York wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I'm worried that your transmission path will also have astigmatism. Even a > perfectly flat piece of glass yields astigmatism in transmission if it's > tilted in a non-infinity space, right? > > (Warning: lousy advice below) > I've encountered this myself, and "fixed" it with a second piece of glass, > the same thickness and material as the first, tilted equally about an > orthogonal axis, to introduce "equal and opposite" astigmatism. This was a > hack, but it was good enough for my needs at the time. Depending on your > needs, perhaps you could use tilted flat glass as a tunable astigmatism > compensation. > > On Mon, Feb 13, 2017 at 8:45 AM, Rusty Nicovich <[hidden email]> > wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Kyle, >> >> The other poster was me. I ultimately fixed the issue to my satisfaction >> by carefully mounting the dichroic to the cube with double-sided tape >> rather than the provided clamp. That method was stress-free enough to not >> induce any undue curvature in the dichroic, at least qualitatively as no >> astigmatism immediately obvious in the reflected image. This was even with >> a relatively thin (1.1 mm) substrate for the mirror. This solution is >> similar to what Jeff suggested using silicone adhesive instead of tape. >> >> Another fun part of that system is that the camera fans would induce quite >> a large vibration and only along one axis. The vibration period was 10-20 >> ms, which meant any longer exposures would yield a PSF elongated on that >> axis. Astigmatism would vary along the focal axis and this didn't, but it >> still gave an asymmetric PSF at the focal plane. >> >> Thanks, >> Rusty >> >> On Mon, Feb 13, 2017 at 7:49 AM, Jeffrey Carmichael < >> [hidden email]> >> wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> Post images on http://www.imgur.com and include the link in your >> posting. >>> ***** >>> >>> COMMERCIAL RESPONSE >>> >>> Hi Kyle, >>> >>> We've found that it's most often the mounting of the dichroic itself >>> causes the most distortion in an image reflected off of a dichroic. You >>> could theoretically have an "ideal" dichroic with no curvature, but once >> it >>> is affixed, it will be torqued out of flatness to some extent unless you >>> can completely avoid any mounting stress. >>> >>> If held by means that apply pressure on top or bottom of the large >> surface, >>> you will get various forms of non-spherical astigmatism. Even light >>> pressure on the sides of the dichroic will have some effect. >>> >>> As Zdenek suggested, thicker dichroics are stiffer and will better resist >>> deformation. >>> >>> If possible, try using something like silicone RTV applied very sparingly >>> in small beads around the outside edges of the dichroic, but never under >>> it. This serves to prevent the dichroic from moving around without >>> applying stress. If the "holder" is removable, then you can swap out >> these >>> mounted dichroics and not fiddle with the dichroic itself once affixed. >>> >>> Jeff >>> >>> >>> *Jeff Carmichael* >>> >>> *Technical and Product Marketing Manager* >>> >>> *[hidden email] <[hidden email]>*Chroma Technology Corp. >>> >>> *an employee owned company* >>> *10 Imtec Lane* >>> *Bellows Falls, VT 05301* >>> *802-428-2528 Office* >>> *802-428-2528 Fax**800-824-7662 Toll Free* >>> >>> On Mon, Feb 13, 2017 at 9:27 AM, Kyle Douglass <[hidden email]> >>> wrote: >>> >>>> ***** >>>> To join, leave or search the confocal microscopy listserv, go to: >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>> Post images on http://www.imgur.com and include the link in your >>> posting. >>>> ***** >>>> >>>> Hi everyone, >>>> >>>> A couple years ago there was a post on this forum about significant >>>> astigmatism appearing in the reflected path of a custom-built, two >> camera >>>> TIRF setup. (http://confocal-microscopy-list.588098.n2.nabble.com/ >> Reflec >>>> ted-image-gt-astigmatism-td7584402.html) The original poster had >> placed >>> a >>>> dichroic before a pair of matched tube lenses to split the two color >>>> channels onto two separate cameras. The general consensus was that you >>>> could "buy your way out of the problem" by buying a thick, ultraflat >>>> dichroic. >>>> >>>> I am trying a similar approach on a setup in our lab but, due to space >>>> constraints that are not easily overcome, have tried first placing the >>>> channel-splitting dichroic in the image space after the final tube lens >>>> instead of in the infinity space between the objective and tube lens. >>> Even >>>> with a 3 mm thick ultraflat dichroic, I see moderately bad astigmatism >> in >>>> the reflected channel. I have ruled out other possible sources of the >>>> astigmatism, like an additional dichroic for the excitation light and >>>> filters that already lie in the infinity space between the objective >> and >>>> tube lens. >>>> >>>> My question is: is it even worth trying to eliminate the astigmatism in >>>> the reflected path of a split-channel setup if the dichroic does not >> lie >>> in >>>> the infinity space? Or is it pretty much always a bad idea to put the >>>> splitter dichroic in the image space? >>>> >>>> My current suspicion is that the tolerances in the alignment have to be >>>> very, very tight to avoid astigmatism with the splitter in the image >>> space. >>>> Thanks for your feedback. >>>> >>>> Kyle >>>> >>>> -- >>>> Kyle M. Douglass, PhD >>>> Post-doctoral researcher >>>> The Laboratory of Experimental Biophysics >>>> EPFL, Lausanne, Switzerland >>>> http://kmdouglass.github.io >>>> http://leb.epfl.ch >>>> -- Kyle M. Douglass, PhD Post-doctoral researcher The Laboratory of Experimental Biophysics EPFL, Lausanne, Switzerland http://kmdouglass.github.io http://leb.epfl.ch "  | 
			
 
	
	 | 
						James Kerin | 
| 
					
		 | 
				
					
	 
				In reply to this post by Andrew York
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** COMMERCIAL RESPONSE Our President and Managing Director have prepared the following response: Hi Kyle Based on our experience of designing image splitters and multicamera adapters, we can strongly underline the points made by Jeff Carmichael. As he says, to avoid any aberrations in the reflected channel, the dichroic must be “softly” mounted against a truly flat surface. Curvature in only one direction will indeed give astigmatism, but more complex curvatures, especially if there is a bit of twist as well, can give significant image distortion too – think fairground mirrors! Even if the images are more or less in focus over the field, this will clearly show up when (as is so often required) the transmitted and reflected images are superimposed for analysis. As camera resolution has improved, such effects have become ever more obvious, but even with megapixel cameras it should be possible to overlay the images to within a just a pixel or two over the field. However that does require similar detail to the design of the rest of the optical system, which should be as symmetrical as possible between the transmitted and reflected pathways, as it is in our image splitters as well as our camera adapters. The flatness issue applies whether the reflection occurs in an infinity space or an imaging space, but there is another problem if the dichroic is in imaging space, namely that of astigmatism in the transmitted channel. This is because the dichroic is at an angle (typically 45 degrees of course) with respect to the light path. Optical modelling shows the following effects, some of which may be counterintuitive. As might be expected though, the astigmatism, which is uniform over the field, increases with the refractive index and thickness of the dichroic, as well as with its angle. (By the way, that's probably why the dichroics used in microscopes to introduce an epi-illumination beam are traditionally “thin”, typically just a millimetre, as in the older (noninfinity) microscopes the dichroic was in the imaging space from the objective). We wonder if this effect might be relevant to Kyle's problem, especially if the dichroic is relatively thick. Relating to that, we have an observation and a suggestion. The optical modelling shows that the astigmatism is independent of the dichroic's position within the imaging space, so moving it closer to either the camera or the focussing lens won't help. We are also pleased from our modelling to confirm that Andrew York's solution to use a second piece of glass mounted at an orthogonal angle is a correct one rather than just a “hack”, even though it may “sound” wrong. The astigmatism arises because the effective thickness of the glass is different in the x and y directions because of it being at an angle. The sign of the angle doesn't matter, so adding another piece of glass angled oppositely would make the problem worse rather than better. But it you rotate by 90 degrees, the directional differences now cancel, and you're just left with a small and probably negligible (according to our modelling at least!) chromatic effect from the dispersion of the glass. Dr MV Thomas & Jez Graham Cairn Research Ltd Graveney Road Faversham Kent, ME13 8UP UK Cairn Research is a European scientific instruments manufacturer based in Kent, UK. www.cairn-research.co.uk On 2/13/2017 9:34 PM, Andrew York wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I'm worried that your transmission path will also have astigmatism. Even a > perfectly flat piece of glass yields astigmatism in transmission if it's > tilted in a non-infinity space, right? > > (Warning: lousy advice below) > I've encountered this myself, and "fixed" it with a second piece of glass, > the same thickness and material as the first, tilted equally about an > orthogonal axis, to introduce "equal and opposite" astigmatism. This was a > hack, but it was good enough for my needs at the time. Depending on your > needs, perhaps you could use tilted flat glass as a tunable astigmatism > compensation. > > On Mon, Feb 13, 2017 at 8:45 AM, Rusty Nicovich <[hidden email]> > wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Kyle, >> >> The other poster was me. I ultimately fixed the issue to my satisfaction >> by carefully mounting the dichroic to the cube with double-sided tape >> rather than the provided clamp. That method was stress-free enough to not >> induce any undue curvature in the dichroic, at least qualitatively as no >> astigmatism immediately obvious in the reflected image. This was even with >> a relatively thin (1.1 mm) substrate for the mirror. This solution is >> similar to what Jeff suggested using silicone adhesive instead of tape. >> >> Another fun part of that system is that the camera fans would induce quite >> a large vibration and only along one axis. The vibration period was 10-20 >> ms, which meant any longer exposures would yield a PSF elongated on that >> axis. Astigmatism would vary along the focal axis and this didn't, but it >> still gave an asymmetric PSF at the focal plane. >> >> Thanks, >> Rusty >> >> On Mon, Feb 13, 2017 at 7:49 AM, Jeffrey Carmichael < >> [hidden email]> >> wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> Post images on http://www.imgur.com and include the link in your >> posting. >>> ***** >>> >>> COMMERCIAL RESPONSE >>> >>> Hi Kyle, >>> >>> We've found that it's most often the mounting of the dichroic itself that >>> causes the most distortion in an image reflected off of a dichroic. You >>> could theoretically have an "ideal" dichroic with no curvature, but once >> it >>> is affixed, it will be torqued out of flatness to some extent unless you >>> can completely avoid any mounting stress. >>> >>> If held by means that apply pressure on top or bottom of the large >> surface, >>> you will get various forms of non-spherical astigmatism. Even light >>> pressure on the sides of the dichroic will have some effect. >>> >>> As Zdenek suggested, thicker dichroics are stiffer and will better resist >>> deformation. >>> >>> If possible, try using something like silicone RTV applied very sparingly >>> in small beads around the outside edges of the dichroic, but never under >>> it. This serves to prevent the dichroic from moving around without >>> applying stress. If the "holder" is removable, then you can swap out >> these >>> mounted dichroics and not fiddle with the dichroic itself once affixed. >>> >>> Jeff >>> >>> >>> *Jeff Carmichael* >>> >>> *Technical and Product Marketing Manager* >>> >>> *[hidden email] <[hidden email]>*Chroma Technology Corp. >>> >>> *an employee owned company* >>> *10 Imtec Lane* >>> *Bellows Falls, VT 05301* >>> *802-428-2528 Office* >>> *802-428-2528 Fax**800-824-7662 Toll Free* >>> >>> On Mon, Feb 13, 2017 at 9:27 AM, Kyle Douglass <[hidden email]> >>> wrote: >>> >>>> ***** >>>> To join, leave or search the confocal microscopy listserv, go to: >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>> Post images on http://www.imgur.com and include the link in your >>> posting. >>>> ***** >>>> >>>> Hi everyone, >>>> >>>> A couple years ago there was a post on this forum about significant >>>> astigmatism appearing in the reflected path of a custom-built, two >> camera >>>> TIRF setup. (http://confocal-microscopy-list.588098.n2.nabble.com/ >> Reflec >>>> ted-image-gt-astigmatism-td7584402.html) The original poster had >> placed >>> a >>>> dichroic before a pair of matched tube lenses to split the two color >>>> channels onto two separate cameras. The general consensus was that you >>>> could "buy your way out of the problem" by buying a thick, ultraflat >>>> dichroic. >>>> >>>> I am trying a similar approach on a setup in our lab but, due to space >>>> constraints that are not easily overcome, have tried first placing the >>>> channel-splitting dichroic in the image space after the final tube lens >>>> instead of in the infinity space between the objective and tube lens. >>> Even >>>> with a 3 mm thick ultraflat dichroic, I see moderately bad astigmatism >> in >>>> the reflected channel. I have ruled out other possible sources of the >>>> astigmatism, like an additional dichroic for the excitation light and >>>> filters that already lie in the infinity space between the objective >> and >>>> tube lens. >>>> >>>> My question is: is it even worth trying to eliminate the astigmatism in >>>> the reflected path of a split-channel setup if the dichroic does not >> lie >>> in >>>> the infinity space? Or is it pretty much always a bad idea to put the >>>> splitter dichroic in the image space? >>>> >>>> My current suspicion is that the tolerances in the alignment have to be >>>> very, very tight to avoid astigmatism with the splitter in the image >>> space. >>>> Thanks for your feedback. >>>> >>>> Kyle >>>> >>>> -- >>>> Kyle M. Douglass, PhD >>>> Post-doctoral researcher >>>> The Laboratory of Experimental Biophysics >>>> EPFL, Lausanne, Switzerland >>>> http://kmdouglass.github.io >>>> http://leb.epfl.ch >>>> -- *James Kerin * *Marketing Director* Cairn Research Ltd Graveney Road Faversham Kent, ME13 8UP UK Direct: + 44 (0)1795 594507 Fax: + 44 (0) 1795 594510  | 
			
 
	
	 | 
						Alfred Millett-Sikking | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Kyle, Our friends and colleagues have done an excellent job of covering this topic, and I very much agree with the contents and summaries provided. I would like to add one further minor point which may or may not be applicable depending on your setup: WFE of any type (including astigmatism) is additive and you mentioned upstream dichroics and filters (and mirrors?). If you're unlucky then it may be that these previous optics have already put you 'on the edge' of an acceptable/perceptible astigmatism and that the last guy is simply pushing you over. If you can't fix it with careful mounting of the last guy then it may be worth giving the upstream optics an equal measure of care (especially if you have reflectors). Without a TWE measurement of the system optic by optic you cannot tell what's going on, so just be suspicious of everyone! Regards, Alfred Millett-Sikking. On 14 February 2017 at 06:35, James Kerin <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > COMMERCIAL RESPONSE > > Our President and Managing Director have prepared the following response: > > Hi Kyle > > Based on our experience of designing image splitters and multicamera > adapters, we can strongly underline the points made by Jeff Carmichael. As > he says, to avoid any aberrations in the reflected channel, the dichroic > must be “softly” mounted against a truly flat surface. Curvature in only > one direction will indeed give astigmatism, but more complex curvatures, > especially if there is a bit of twist as well, can give significant image > distortion too – think fairground mirrors! Even if the images are more or > less in focus over the field, this will clearly show up when (as is so > often required) the transmitted and reflected images are superimposed for > analysis. As camera resolution has improved, such effects have become ever > more obvious, but even with megapixel cameras it should be possible to > overlay the images to within a just a pixel or two over the field. However > that does require similar detail to the design of the rest of the optical > system, which should be as symmetrical as possible between the transmitted > and reflected pathways, as it is in our image splitters as well as our > camera adapters. > > The flatness issue applies whether the reflection occurs in an infinity > space or an imaging space, but there is another problem if the dichroic is > in imaging space, namely that of astigmatism in the transmitted channel. > This is because the dichroic is at an angle (typically 45 degrees of > course) with respect to the light path. Optical modelling shows the > following effects, some of which may be counterintuitive. As might be > expected though, the astigmatism, which is uniform over the field, > increases with the refractive index and thickness of the dichroic, as well > as with its angle. (By the way, that's probably why the dichroics used in > microscopes to introduce an epi-illumination beam are traditionally “thin”, > typically just a millimetre, as in the older (noninfinity) microscopes the > dichroic was in the imaging space from the objective). > > We wonder if this effect might be relevant to Kyle's problem, especially > if the dichroic is relatively thick. Relating to that, we have an > observation and a suggestion. The optical modelling shows that the > astigmatism is independent of the dichroic's position within the imaging > space, so moving it closer to either the camera or the focussing lens won't > help. > > We are also pleased from our modelling to confirm that Andrew York's > solution to use a second piece of glass mounted at an orthogonal angle is a > correct one rather than just a “hack”, even though it may “sound” wrong. > The astigmatism arises because the effective thickness of the glass is > different in the x and y directions because of it being at an angle. The > sign of the angle doesn't matter, so adding another piece of glass angled > oppositely would make the problem worse rather than better. But it you > rotate by 90 degrees, the directional differences now cancel, and you're > just left with a small and probably negligible (according to our modelling > at least!) chromatic effect from the dispersion of the glass. > > Dr MV Thomas & Jez Graham > Cairn Research Ltd > Graveney Road > Faversham > Kent, ME13 8UP > UK > Cairn Research is a European scientific instruments manufacturer based in > Kent, UK. > www.cairn-research.co.uk > > > > > > On 2/13/2017 9:34 PM, Andrew York wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> I'm worried that your transmission path will also have astigmatism. >> Even a >> perfectly flat piece of glass yields astigmatism in transmission if it's >> tilted in a non-infinity space, right? >> >> (Warning: lousy advice below) >> I've encountered this myself, and "fixed" it with a second piece of >> glass, >> the same thickness and material as the first, tilted equally about an >> orthogonal axis, to introduce "equal and opposite" astigmatism. This was a >> hack, but it was good enough for my needs at the time. Depending on your >> needs, perhaps you could use tilted flat glass as a tunable astigmatism >> compensation. >> >> On Mon, Feb 13, 2017 at 8:45 AM, Rusty Nicovich <[hidden email]> >> wrote: >> >> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> Post images on http://www.imgur.com and include the link in your >>> posting. >>> ***** >>> >>> Kyle, >>> >>> The other poster was me. I ultimately fixed the issue to my satisfaction >>> by carefully mounting the dichroic to the cube with double-sided tape >>> rather than the provided clamp. That method was stress-free enough to >>> not >>> induce any undue curvature in the dichroic, at least qualitatively as no >>> astigmatism immediately obvious in the reflected image. This was even >>> with >>> a relatively thin (1.1 mm) substrate for the mirror. This solution is >>> similar to what Jeff suggested using silicone adhesive instead of tape. >>> >>> Another fun part of that system is that the camera fans would induce >>> quite >>> a large vibration and only along one axis. The vibration period was >>> 10-20 >>> ms, which meant any longer exposures would yield a PSF elongated on that >>> axis. Astigmatism would vary along the focal axis and this didn't, but >>> it >>> still gave an asymmetric PSF at the focal plane. >>> >>> Thanks, >>> Rusty >>> >>> On Mon, Feb 13, 2017 at 7:49 AM, Jeffrey Carmichael < >>> [hidden email]> >>> wrote: >>> >>> ***** >>>> To join, leave or search the confocal microscopy listserv, go to: >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>> Post images on http://www.imgur.com and include the link in your >>>> >>> posting. >>> >>>> ***** >>>> >>>> COMMERCIAL RESPONSE >>>> >>>> Hi Kyle, >>>> >>>> We've found that it's most often the mounting of the dichroic itself >>>> that >>>> causes the most distortion in an image reflected off of a dichroic. You >>>> could theoretically have an "ideal" dichroic with no curvature, but once >>>> >>> it >>> >>>> is affixed, it will be torqued out of flatness to some extent unless you >>>> can completely avoid any mounting stress. >>>> >>>> If held by means that apply pressure on top or bottom of the large >>>> >>> surface, >>> >>>> you will get various forms of non-spherical astigmatism. Even light >>>> pressure on the sides of the dichroic will have some effect. >>>> >>>> As Zdenek suggested, thicker dichroics are stiffer and will better >>>> resist >>>> deformation. >>>> >>>> If possible, try using something like silicone RTV applied very >>>> sparingly >>>> in small beads around the outside edges of the dichroic, but never under >>>> it. This serves to prevent the dichroic from moving around without >>>> applying stress. If the "holder" is removable, then you can swap out >>>> >>> these >>> >>>> mounted dichroics and not fiddle with the dichroic itself once affixed. >>>> >>>> Jeff >>>> >>>> >>>> *Jeff Carmichael* >>>> >>>> *Technical and Product Marketing Manager* >>>> >>>> *[hidden email] <[hidden email]>*Chroma Technology >>>> Corp. >>>> >>>> *an employee owned company* >>>> *10 Imtec Lane* >>>> *Bellows Falls, VT 05301* >>>> *802-428-2528 Office* >>>> *802-428-2528 Fax**800-824-7662 Toll Free* >>>> >>>> On Mon, Feb 13, 2017 at 9:27 AM, Kyle Douglass <[hidden email]> >>>> wrote: >>>> >>>> ***** >>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>> Post images on http://www.imgur.com and include the link in your >>>>> >>>> posting. >>>> >>>>> ***** >>>>> >>>>> Hi everyone, >>>>> >>>>> A couple years ago there was a post on this forum about significant >>>>> astigmatism appearing in the reflected path of a custom-built, two >>>>> >>>> camera >>> >>>> TIRF setup. (http://confocal-microscopy-list.588098.n2.nabble.com/ >>>>> >>>> Reflec >>> >>>> ted-image-gt-astigmatism-td7584402.html) The original poster had >>>>> >>>> placed >>> >>>> a >>>> >>>>> dichroic before a pair of matched tube lenses to split the two color >>>>> channels onto two separate cameras. The general consensus was that you >>>>> could "buy your way out of the problem" by buying a thick, ultraflat >>>>> dichroic. >>>>> >>>>> I am trying a similar approach on a setup in our lab but, due to space >>>>> constraints that are not easily overcome, have tried first placing the >>>>> channel-splitting dichroic in the image space after the final tube lens >>>>> instead of in the infinity space between the objective and tube lens. >>>>> >>>> Even >>>> >>>>> with a 3 mm thick ultraflat dichroic, I see moderately bad astigmatism >>>>> >>>> in >>> >>>> the reflected channel. I have ruled out other possible sources of the >>>>> astigmatism, like an additional dichroic for the excitation light and >>>>> filters that already lie in the infinity space between the objective >>>>> >>>> and >>> >>>> tube lens. >>>>> >>>>> My question is: is it even worth trying to eliminate the astigmatism in >>>>> the reflected path of a split-channel setup if the dichroic does not >>>>> >>>> lie >>> >>>> in >>>> >>>>> the infinity space? Or is it pretty much always a bad idea to put the >>>>> splitter dichroic in the image space? >>>>> >>>>> My current suspicion is that the tolerances in the alignment have to be >>>>> very, very tight to avoid astigmatism with the splitter in the image >>>>> >>>> space. >>>> >>>>> Thanks for your feedback. >>>>> >>>>> Kyle >>>>> >>>>> -- >>>>> Kyle M. Douglass, PhD >>>>> Post-doctoral researcher >>>>> The Laboratory of Experimental Biophysics >>>>> EPFL, Lausanne, Switzerland >>>>> http://kmdouglass.github.io >>>>> http://leb.epfl.ch >>>>> >>>>> > -- > > *James Kerin > * > > *Marketing Director* > Cairn Research Ltd > Graveney Road > Faversham > Kent, ME13 8UP > UK > > Direct: + 44 (0)1795 594507 > Fax: + 44 (0) 1795 594510 >  | 
			
 
	
	 | 
						Andrew York | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Is the reflective side of the dichroic facing the incoming light? If not, you might get astigmatism in reflection even from a perfectly flat dichroic in a non-infinity space. On Tue, Feb 14, 2017 at 10:51 AM, Alfred Millett-Sikking < [hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Kyle, > > Our friends and colleagues have done an excellent job of covering this > topic, and I very much agree with the contents and summaries provided. I > would like to add one further minor point which may or may not be > applicable depending on your setup: > > WFE of any type (including astigmatism) is additive and you mentioned > upstream dichroics and filters (and mirrors?). If you're unlucky then it > may be that these previous optics have already put you 'on the edge' of an > acceptable/perceptible astigmatism and that the last guy is simply pushing > you over. If you can't fix it with careful mounting of the last guy then it > may be worth giving the upstream optics an equal measure of care > (especially if you have reflectors). Without a TWE measurement of the > system optic by optic you cannot tell what's going on, so just be > suspicious of everyone! > > Regards, > > Alfred Millett-Sikking. > > > On 14 February 2017 at 06:35, James Kerin <[hidden email]> > wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > COMMERCIAL RESPONSE > > > > Our President and Managing Director have prepared the following response: > > > > Hi Kyle > > > > Based on our experience of designing image splitters and multicamera > > adapters, we can strongly underline the points made by Jeff Carmichael. > As > > he says, to avoid any aberrations in the reflected channel, the dichroic > > must be “softly” mounted against a truly flat surface. Curvature in only > > one direction will indeed give astigmatism, but more complex curvatures, > > especially if there is a bit of twist as well, can give significant image > > distortion too – think fairground mirrors! Even if the images are more > or > > less in focus over the field, this will clearly show up when (as is so > > often required) the transmitted and reflected images are superimposed for > > analysis. As camera resolution has improved, such effects have become > ever > > more obvious, but even with megapixel cameras it should be possible to > > overlay the images to within a just a pixel or two over the field. > However > > that does require similar detail to the design of the rest of the optical > > system, which should be as symmetrical as possible between the > transmitted > > and reflected pathways, as it is in our image splitters as well as our > > camera adapters. > > > > The flatness issue applies whether the reflection occurs in an infinity > > space or an imaging space, but there is another problem if the dichroic > is > > in imaging space, namely that of astigmatism in the transmitted channel. > > This is because the dichroic is at an angle (typically 45 degrees of > > course) with respect to the light path. Optical modelling shows the > > following effects, some of which may be counterintuitive. As might be > > expected though, the astigmatism, which is uniform over the field, > > increases with the refractive index and thickness of the dichroic, as > well > > as with its angle. (By the way, that's probably why the dichroics used > in > > microscopes to introduce an epi-illumination beam are traditionally > “thin”, > > typically just a millimetre, as in the older (noninfinity) microscopes > the > > dichroic was in the imaging space from the objective). > > > > We wonder if this effect might be relevant to Kyle's problem, especially > > if the dichroic is relatively thick. Relating to that, we have an > > observation and a suggestion. The optical modelling shows that the > > astigmatism is independent of the dichroic's position within the imaging > > space, so moving it closer to either the camera or the focussing lens > won't > > help. > > > > We are also pleased from our modelling to confirm that Andrew York's > > solution to use a second piece of glass mounted at an orthogonal angle > is a > > correct one rather than just a “hack”, even though it may “sound” wrong. > > The astigmatism arises because the effective thickness of the glass is > > different in the x and y directions because of it being at an angle. The > > sign of the angle doesn't matter, so adding another piece of glass angled > > oppositely would make the problem worse rather than better. But it you > > rotate by 90 degrees, the directional differences now cancel, and you're > > just left with a small and probably negligible (according to our > modelling > > at least!) chromatic effect from the dispersion of the glass. > > > > Dr MV Thomas & Jez Graham > > Cairn Research Ltd > > Graveney Road > > Faversham > > Kent, ME13 8UP > > UK > > Cairn Research is a European scientific instruments manufacturer based in > > Kent, UK. > > www.cairn-research.co.uk > > > > > > > > > > > > On 2/13/2017 9:34 PM, Andrew York wrote: > > > >> ***** > >> To join, leave or search the confocal microscopy listserv, go to: > >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >> Post images on http://www.imgur.com and include the link in your > posting. > >> ***** > >> > >> I'm worried that your transmission path will also have astigmatism. > >> Even a > >> perfectly flat piece of glass yields astigmatism in transmission if it's > >> tilted in a non-infinity space, right? > >> > >> (Warning: lousy advice below) > >> I've encountered this myself, and "fixed" it with a second piece of > >> glass, > >> the same thickness and material as the first, tilted equally about an > >> orthogonal axis, to introduce "equal and opposite" astigmatism. This > was a > >> hack, but it was good enough for my needs at the time. Depending on your > >> needs, perhaps you could use tilted flat glass as a tunable astigmatism > >> compensation. > >> > >> On Mon, Feb 13, 2017 at 8:45 AM, Rusty Nicovich < > [hidden email]> > >> wrote: > >> > >> ***** > >>> To join, leave or search the confocal microscopy listserv, go to: > >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >>> Post images on http://www.imgur.com and include the link in your > >>> posting. > >>> ***** > >>> > >>> Kyle, > >>> > >>> The other poster was me. I ultimately fixed the issue to my > satisfaction > >>> by carefully mounting the dichroic to the cube with double-sided tape > >>> rather than the provided clamp. That method was stress-free enough to > >>> not > >>> induce any undue curvature in the dichroic, at least qualitatively as > no > >>> astigmatism immediately obvious in the reflected image. This was even > >>> with > >>> a relatively thin (1.1 mm) substrate for the mirror. This solution is > >>> similar to what Jeff suggested using silicone adhesive instead of tape. > >>> > >>> Another fun part of that system is that the camera fans would induce > >>> quite > >>> a large vibration and only along one axis. The vibration period was > >>> 10-20 > >>> ms, which meant any longer exposures would yield a PSF elongated on > that > >>> axis. Astigmatism would vary along the focal axis and this didn't, but > >>> it > >>> still gave an asymmetric PSF at the focal plane. > >>> > >>> Thanks, > >>> Rusty > >>> > >>> On Mon, Feb 13, 2017 at 7:49 AM, Jeffrey Carmichael < > >>> [hidden email]> > >>> wrote: > >>> > >>> ***** > >>>> To join, leave or search the confocal microscopy listserv, go to: > >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >>>> Post images on http://www.imgur.com and include the link in your > >>>> > >>> posting. > >>> > >>>> ***** > >>>> > >>>> COMMERCIAL RESPONSE > >>>> > >>>> Hi Kyle, > >>>> > >>>> We've found that it's most often the mounting of the dichroic itself > >>>> that > >>>> causes the most distortion in an image reflected off of a dichroic. > You > >>>> could theoretically have an "ideal" dichroic with no curvature, but > once > >>>> > >>> it > >>> > >>>> is affixed, it will be torqued out of flatness to some extent unless > you > >>>> can completely avoid any mounting stress. > >>>> > >>>> If held by means that apply pressure on top or bottom of the large > >>>> > >>> surface, > >>> > >>>> you will get various forms of non-spherical astigmatism. Even light > >>>> pressure on the sides of the dichroic will have some effect. > >>>> > >>>> As Zdenek suggested, thicker dichroics are stiffer and will better > >>>> resist > >>>> deformation. > >>>> > >>>> If possible, try using something like silicone RTV applied very > >>>> sparingly > >>>> in small beads around the outside edges of the dichroic, but never > under > >>>> it. This serves to prevent the dichroic from moving around without > >>>> applying stress. If the "holder" is removable, then you can swap out > >>>> > >>> these > >>> > >>>> mounted dichroics and not fiddle with the dichroic itself once > affixed. > >>>> > >>>> Jeff > >>>> > >>>> > >>>> *Jeff Carmichael* > >>>> > >>>> *Technical and Product Marketing Manager* > >>>> > >>>> *[hidden email] <[hidden email]>*Chroma Technology > >>>> Corp. > >>>> > >>>> *an employee owned company* > >>>> *10 Imtec Lane* > >>>> *Bellows Falls, VT 05301* > >>>> *802-428-2528 Office* > >>>> *802-428-2528 Fax**800-824-7662 Toll Free* > >>>> > >>>> On Mon, Feb 13, 2017 at 9:27 AM, Kyle Douglass <[hidden email] > > > >>>> wrote: > >>>> > >>>> ***** > >>>>> To join, leave or search the confocal microscopy listserv, go to: > >>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >>>>> Post images on http://www.imgur.com and include the link in your > >>>>> > >>>> posting. > >>>> > >>>>> ***** > >>>>> > >>>>> Hi everyone, > >>>>> > >>>>> A couple years ago there was a post on this forum about significant > >>>>> astigmatism appearing in the reflected path of a custom-built, two > >>>>> > >>>> camera > >>> > >>>> TIRF setup. (http://confocal-microscopy-list.588098.n2.nabble.com/ > >>>>> > >>>> Reflec > >>> > >>>> ted-image-gt-astigmatism-td7584402.html) The original poster had > >>>>> > >>>> placed > >>> > >>>> a > >>>> > >>>>> dichroic before a pair of matched tube lenses to split the two color > >>>>> channels onto two separate cameras. The general consensus was that > you > >>>>> could "buy your way out of the problem" by buying a thick, ultraflat > >>>>> dichroic. > >>>>> > >>>>> I am trying a similar approach on a setup in our lab but, due to > space > >>>>> constraints that are not easily overcome, have tried first placing > the > >>>>> channel-splitting dichroic in the image space after the final tube > lens > >>>>> instead of in the infinity space between the objective and tube lens. > >>>>> > >>>> Even > >>>> > >>>>> with a 3 mm thick ultraflat dichroic, I see moderately bad > astigmatism > >>>>> > >>>> in > >>> > >>>> the reflected channel. I have ruled out other possible sources of the > >>>>> astigmatism, like an additional dichroic for the excitation light and > >>>>> filters that already lie in the infinity space between the objective > >>>>> > >>>> and > >>> > >>>> tube lens. > >>>>> > >>>>> My question is: is it even worth trying to eliminate the astigmatism > in > >>>>> the reflected path of a split-channel setup if the dichroic does not > >>>>> > >>>> lie > >>> > >>>> in > >>>> > >>>>> the infinity space? Or is it pretty much always a bad idea to put the > >>>>> splitter dichroic in the image space? > >>>>> > >>>>> My current suspicion is that the tolerances in the alignment have to > be > >>>>> very, very tight to avoid astigmatism with the splitter in the image > >>>>> > >>>> space. > >>>> > >>>>> Thanks for your feedback. > >>>>> > >>>>> Kyle > >>>>> > >>>>> -- > >>>>> Kyle M. Douglass, PhD > >>>>> Post-doctoral researcher > >>>>> The Laboratory of Experimental Biophysics > >>>>> EPFL, Lausanne, Switzerland > >>>>> http://kmdouglass.github.io > >>>>> http://leb.epfl.ch > >>>>> > >>>>> > > -- > > > > *James Kerin > > * > > > > *Marketing Director* > > Cairn Research Ltd > > Graveney Road > > Faversham > > Kent, ME13 8UP > > UK > > > > Direct: + 44 (0)1795 594507 > > Fax: + 44 (0) 1795 594510 > > >  | 
			
 
	
	 | 
						Kyle Michael Douglass | 
| 
					
		 | 
				
					
	 
				In reply to this post by Alfred Millett-Sikking
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thanks everyone for your replies. I should have enough info to better troubleshoot the problem. If I manage to find a solution, I'll post it to this same thread so future listers can benefit. Cheers, Kyle On 02/14/2017 07:51 PM, Alfred Millett-Sikking wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Kyle, > > Our friends and colleagues have done an excellent job of covering this > topic, and I very much agree with the contents and summaries provided. I > would like to add one further minor point which may or may not be > applicable depending on your setup: > > WFE of any type (including astigmatism) is additive and you mentioned > upstream dichroics and filters (and mirrors?). If you're unlucky then it > may be that these previous optics have already put you 'on the edge' of an > acceptable/perceptible astigmatism and that the last guy is simply pushing > you over. If you can't fix it with careful mounting of the last guy then it > may be worth giving the upstream optics an equal measure of care > (especially if you have reflectors). Without a TWE measurement of the > system optic by optic you cannot tell what's going on, so just be > suspicious of everyone! > > Regards, > > Alfred Millett-Sikking. > > > On 14 February 2017 at 06:35, James Kerin <[hidden email]> > wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> COMMERCIAL RESPONSE >> >> Our President and Managing Director have prepared the following response: >> >> Hi Kyle >> >> Based on our experience of designing image splitters and multicamera >> adapters, we can strongly underline the points made by Jeff Carmichael. As >> he says, to avoid any aberrations in the reflected channel, the dichroic >> must be “softly” mounted against a truly flat surface. Curvature in only >> one direction will indeed give astigmatism, but more complex curvatures, >> especially if there is a bit of twist as well, can give significant image >> distortion too – think fairground mirrors! Even if the images are more or >> less in focus over the field, this will clearly show up when (as is so >> often required) the transmitted and reflected images are superimposed for >> analysis. As camera resolution has improved, such effects have become ever >> more obvious, but even with megapixel cameras it should be possible to >> overlay the images to within a just a pixel or two over the field. However >> that does require similar detail to the design of the rest of the optical >> system, which should be as symmetrical as possible between the transmitted >> and reflected pathways, as it is in our image splitters as well as our >> camera adapters. >> >> The flatness issue applies whether the reflection occurs in an infinity >> space or an imaging space, but there is another problem if the dichroic is >> in imaging space, namely that of astigmatism in the transmitted channel. >> This is because the dichroic is at an angle (typically 45 degrees of >> course) with respect to the light path. Optical modelling shows the >> following effects, some of which may be counterintuitive. As might be >> expected though, the astigmatism, which is uniform over the field, >> increases with the refractive index and thickness of the dichroic, as well >> as with its angle. (By the way, that's probably why the dichroics used in >> microscopes to introduce an epi-illumination beam are traditionally “thin”, >> typically just a millimetre, as in the older (noninfinity) microscopes the >> dichroic was in the imaging space from the objective). >> >> We wonder if this effect might be relevant to Kyle's problem, especially >> if the dichroic is relatively thick. Relating to that, we have an >> observation and a suggestion. The optical modelling shows that the >> astigmatism is independent of the dichroic's position within the imaging >> space, so moving it closer to either the camera or the focussing lens won't >> help. >> >> We are also pleased from our modelling to confirm that Andrew York's >> solution to use a second piece of glass mounted at an orthogonal angle is a >> correct one rather than just a “hack”, even though it may “sound” wrong. >> The astigmatism arises because the effective thickness of the glass is >> different in the x and y directions because of it being at an angle. The >> sign of the angle doesn't matter, so adding another piece of glass angled >> oppositely would make the problem worse rather than better. But it you >> rotate by 90 degrees, the directional differences now cancel, and you're >> just left with a small and probably negligible (according to our modelling >> at least!) chromatic effect from the dispersion of the glass. >> >> Dr MV Thomas & Jez Graham >> Cairn Research Ltd >> Graveney Road >> Faversham >> Kent, ME13 8UP >> UK >> Cairn Research is a European scientific instruments manufacturer based in >> Kent, UK. >> www.cairn-research.co.uk >> >> >> >> >> >> On 2/13/2017 9:34 PM, Andrew York wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> Post images on http://www.imgur.com and include the link in your posting. >>> ***** >>> >>> I'm worried that your transmission path will also have astigmatism. >>> Even a >>> perfectly flat piece of glass yields astigmatism in transmission if it's >>> tilted in a non-infinity space, right? >>> >>> (Warning: lousy advice below) >>> I've encountered this myself, and "fixed" it with a second piece of >>> glass, >>> the same thickness and material as the first, tilted equally about an >>> orthogonal axis, to introduce "equal and opposite" astigmatism. This was a >>> hack, but it was good enough for my needs at the time. Depending on your >>> needs, perhaps you could use tilted flat glass as a tunable astigmatism >>> compensation. >>> >>> On Mon, Feb 13, 2017 at 8:45 AM, Rusty Nicovich <[hidden email]> >>> wrote: >>> >>> ***** >>>> To join, leave or search the confocal microscopy listserv, go to: >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>> Post images on http://www.imgur.com and include the link in your >>>> posting. >>>> ***** >>>> >>>> Kyle, >>>> >>>> The other poster was me. I ultimately fixed the issue to my satisfaction >>>> by carefully mounting the dichroic to the cube with double-sided tape >>>> rather than the provided clamp. That method was stress-free enough to >>>> not >>>> induce any undue curvature in the dichroic, at least qualitatively as no >>>> astigmatism immediately obvious in the reflected image. This was even >>>> with >>>> a relatively thin (1.1 mm) substrate for the mirror. This solution is >>>> similar to what Jeff suggested using silicone adhesive instead of tape. >>>> >>>> Another fun part of that system is that the camera fans would induce >>>> quite >>>> a large vibration and only along one axis. The vibration period was >>>> 10-20 >>>> ms, which meant any longer exposures would yield a PSF elongated on that >>>> axis. Astigmatism would vary along the focal axis and this didn't, but >>>> it >>>> still gave an asymmetric PSF at the focal plane. >>>> >>>> Thanks, >>>> Rusty >>>> >>>> On Mon, Feb 13, 2017 at 7:49 AM, Jeffrey Carmichael < >>>> [hidden email]> >>>> wrote: >>>> >>>> ***** >>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>> Post images on http://www.imgur.com and include the link in your >>>>> >>>> posting. >>>> >>>>> ***** >>>>> >>>>> COMMERCIAL RESPONSE >>>>> >>>>> Hi Kyle, >>>>> >>>>> We've found that it's most often the mounting of the dichroic itself >>>>> that >>>>> causes the most distortion in an image reflected off of a dichroic. You >>>>> could theoretically have an "ideal" dichroic with no curvature, but once >>>>> >>>> it >>>> >>>>> is affixed, it will be torqued out of flatness to some extent unless you >>>>> can completely avoid any mounting stress. >>>>> >>>>> If held by means that apply pressure on top or bottom of the large >>>>> >>>> surface, >>>> >>>>> you will get various forms of non-spherical astigmatism. Even light >>>>> pressure on the sides of the dichroic will have some effect. >>>>> >>>>> As Zdenek suggested, thicker dichroics are stiffer and will better >>>>> resist >>>>> deformation. >>>>> >>>>> If possible, try using something like silicone RTV applied very >>>>> sparingly >>>>> in small beads around the outside edges of the dichroic, but never under >>>>> it. This serves to prevent the dichroic from moving around without >>>>> applying stress. If the "holder" is removable, then you can swap out >>>>> >>>> these >>>> >>>>> mounted dichroics and not fiddle with the dichroic itself once affixed. >>>>> >>>>> Jeff >>>>> >>>>> >>>>> *Jeff Carmichael* >>>>> >>>>> *Technical and Product Marketing Manager* >>>>> >>>>> *[hidden email] <[hidden email]>*Chroma Technology >>>>> Corp. >>>>> >>>>> *an employee owned company* >>>>> *10 Imtec Lane* >>>>> *Bellows Falls, VT 05301* >>>>> *802-428-2528 Office* >>>>> *802-428-2528 Fax**800-824-7662 Toll Free* >>>>> >>>>> On Mon, Feb 13, 2017 at 9:27 AM, Kyle Douglass <[hidden email]> >>>>> wrote: >>>>> >>>>> ***** >>>>>> To join, leave or search the confocal microscopy listserv, go to: >>>>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>>>> Post images on http://www.imgur.com and include the link in your >>>>>> >>>>> posting. >>>>> >>>>>> ***** >>>>>> >>>>>> Hi everyone, >>>>>> >>>>>> A couple years ago there was a post on this forum about significant >>>>>> astigmatism appearing in the reflected path of a custom-built, two >>>>>> >>>>> camera >>>>> TIRF setup. (http://confocal-microscopy-list.588098.n2.nabble.com/ >>>>> Reflec >>>>> ted-image-gt-astigmatism-td7584402.html) The original poster had >>>>> placed >>>>> a >>>>> >>>>>> dichroic before a pair of matched tube lenses to split the two color >>>>>> channels onto two separate cameras. The general consensus was that you >>>>>> could "buy your way out of the problem" by buying a thick, ultraflat >>>>>> dichroic. >>>>>> >>>>>> I am trying a similar approach on a setup in our lab but, due to space >>>>>> constraints that are not easily overcome, have tried first placing the >>>>>> channel-splitting dichroic in the image space after the final tube lens >>>>>> instead of in the infinity space between the objective and tube lens. >>>>>> >>>>> Even >>>>> >>>>>> with a 3 mm thick ultraflat dichroic, I see moderately bad astigmatism >>>>>> >>>>> in >>>>> the reflected channel. I have ruled out other possible sources of the >>>>>> astigmatism, like an additional dichroic for the excitation light and >>>>>> filters that already lie in the infinity space between the objective >>>>>> >>>>> and >>>>> tube lens. >>>>>> My question is: is it even worth trying to eliminate the astigmatism in >>>>>> the reflected path of a split-channel setup if the dichroic does not >>>>>> >>>>> lie >>>>> in >>>>> >>>>>> the infinity space? Or is it pretty much always a bad idea to put the >>>>>> splitter dichroic in the image space? >>>>>> >>>>>> My current suspicion is that the tolerances in the alignment have to be >>>>>> very, very tight to avoid astigmatism with the splitter in the image >>>>>> >>>>> space. >>>>> >>>>>> Thanks for your feedback. >>>>>> >>>>>> Kyle >>>>>> >>>>>> -- >>>>>> Kyle M. Douglass, PhD >>>>>> Post-doctoral researcher >>>>>> The Laboratory of Experimental Biophysics >>>>>> EPFL, Lausanne, Switzerland >>>>>> http://kmdouglass.github.io >>>>>> http://leb.epfl.ch >>>>>> >>>>>> >> -- >> >> *James Kerin >> * >> >> *Marketing Director* >> Cairn Research Ltd >> Graveney Road >> Faversham >> Kent, ME13 8UP >> UK >> >> Direct: + 44 (0)1795 594507 >> Fax: + 44 (0) 1795 594510 >> -- Kyle M. Douglass, PhD Post-doctoral researcher The Laboratory of Experimental Biophysics EPFL, Lausanne, Switzerland http://kmdouglass.github.io http://leb.epfl.ch  | 
			