				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear all, For a couple of months we are struggling with a serious problem with the field illumination in our Olympus Scan^R Scanning station. The image of the specimen, when observed on the computer screen is much darker on one side than on the other. We thought it is the lamp but the lamp is new at properly inserted. Then we tried calibrating the optic fiber, but it didn't help at all. Our third guess was the filters.We checked all of the filters - the coating has lots of small dark spots with biggest concentration of the spots in the middle of the filters, but I am concerned if that can produce uneven illumination of the field so that only one part (let say left corner) is underilluminated? So, I turn to you and your experience: did may anyone have the similar problem? Do you have any suggestion what can be the cause of the problem? Best regards, Hanna -- Department of Plant Anatomy and Cytology Faculty of Biology and Environmental Protection University of Silesia Jagiellonska Str. 28 40-032 Katowice Poland  | 
			
 
	
	 | 
						tineke vendrig | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Have you optimised you microscope ?? Kohler illumination and so on..?? Tineke Vendrig, ing technical engeneer optical microscopy Delft University of Technology Bionano Science Kavli Institute of Nanoscience Lorentzweg 1 2628LJ Delft room F185 Tel: +31 27 89299 Fax:+31 15 2781202 email: [hidden email] mobile phone: 06-34241412 2011/10/6 Hanna Sas Nowosielska <[hidden email]> > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all, > > For a couple of months we are struggling with a serious problem with > the field illumination in our Olympus Scan^R Scanning station. The > image of the specimen, when observed on the computer screen is much > darker on one side than on the other. We thought it is the lamp but > the lamp is new at properly inserted. Then we tried calibrating the > optic fiber, but it didn't help at all. Our third guess was the > filters.We checked all of the filters - the coating has lots of > small dark spots with biggest concentration of the spots in the middle > of the filters, but I am concerned if that can produce uneven > illumination of the field so that only one part (let say left corner) > is underilluminated? So, I turn to you and your experience: did may > anyone have the similar problem? Do you have any suggestion what can > be the cause of the problem? > > Best regards, > Hanna > > -- > Department of Plant Anatomy and Cytology > Faculty of Biology and Environmental Protection > University of Silesia > Jagiellonska Str. 28 > 40-032 Katowice > Poland > -- Tineke Vendrig, ing technical engeneer optical microscopy Delft University of Technology Bionano Science Kavli Institute of Nanoscience Lorentzweg 1 2628LJ Delft room F185 Tel: +31 27 89299 Fax:+31 15 2781202 email: [hidden email] mobile phone: 06-34241412  | 
			
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Yes, at least Olymus claims everything with the optical path is ok. Hanna 2011/10/6 tineke vendrig <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Have you optimised you microscope ?? Kohler illumination and so on..?? > > Tineke Vendrig, ing > technical engeneer optical microscopy > Delft University of Technology > Bionano Science > Kavli Institute of Nanoscience > Lorentzweg 1 > 2628LJ Delft > room F185 > Tel: +31 27 89299 > Fax:+31 15 2781202 > email: [hidden email] > mobile phone: 06-34241412 > > > 2011/10/6 Hanna Sas Nowosielska <[hidden email]> > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear all, >> >> For a couple of months we are struggling with a serious problem with >> the field illumination in our Olympus Scan^R Scanning station. The >> image of the specimen, when observed on the computer screen is much >> darker on one side than on the other. We thought it is the lamp but >> the lamp is new at properly inserted. Then we tried calibrating the >> optic fiber, but it didn't help at all. Our third guess was the >> filters.We checked all of the filters - the coating has lots of >> small dark spots with biggest concentration of the spots in the middle >> of the filters, but I am concerned if that can produce uneven >> illumination of the field so that only one part (let say left corner) >> is underilluminated? So, I turn to you and your experience: did may >> anyone have the similar problem? Do you have any suggestion what can >> be the cause of the problem? >> >> Best regards, >> Hanna >> >> -- >> Department of Plant Anatomy and Cytology >> Faculty of Biology and Environmental Protection >> University of Silesia >> Jagiellonska Str. 28 >> 40-032 Katowice >> Poland >> > > > > -- > Tineke Vendrig, ing > technical engeneer optical microscopy > Delft University of Technology > Bionano Science > Kavli Institute of Nanoscience > Lorentzweg 1 > 2628LJ Delft > room F185 > Tel: +31 27 89299 > Fax:+31 15 2781202 > email: [hidden email] > mobile phone: 06-34241412 > -- Department of Plant Anatomy and Cytology Faculty of Biology and Environmental Protection University of Silesia Jagiellonska Str. 28 40-032 Katowice Poland  | 
			
 
	
	 | 
						Arne Seitz | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** If the beam path is ok then you do not have uneven illumination. Or the other way round: Uneven illumination is a perfect sign that the beam path is not ok (but a service technician should be able to fix it). Cheers Arne > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] On Behalf Of Hanna Sas > Nowosielska > Sent: jeudi 6 octobre 2011 16:37 > To: [hidden email] > Subject: Re: Rejected posting to [hidden email] > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Yes, at least Olymus claims everything with the optical path is ok. > > Hanna > > 2011/10/6 tineke vendrig <[hidden email]>: > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Have you optimised you microscope ?? Kohler illumination and so on..?? > > > > Tineke Vendrig, ing > > technical engeneer optical microscopy > > Delft University of Technology > > Bionano Science > > Kavli Institute of Nanoscience > > Lorentzweg 1 > > 2628LJ Delft > > room F185 > > Tel: +31 27 89299 > > Fax:+31 15 2781202 > > email: [hidden email] > > mobile phone: 06-34241412 > > > > > > 2011/10/6 Hanna Sas Nowosielska <[hidden email]> > > > >> ***** > >> To join, leave or search the confocal microscopy listserv, go to: > >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >> ***** > >> > >> Dear all, > >> > >> For a couple of months we are struggling with a serious problem with > >> the field illumination in our Olympus Scan^R Scanning station. The > >> image of the specimen, when observed on the computer screen is much > >> darker on one side than on the other. We thought it is the lamp but > >> the lamp is new at properly inserted. Then we tried calibrating the > >> optic fiber, but it didn't help at all. Our third guess was the > >> filters.We checked all of the filters - the coating has lots of > >> small dark spots with biggest concentration of the spots in the > >> middle of the filters, but I am concerned if that can produce uneven > >> illumination of the field so that only one part (let say left corner) > >> is underilluminated? So, I turn to you and your experience: did may > >> anyone have the similar problem? Do you have any suggestion what can > >> be the cause of the problem? > >> > >> Best regards, > >> Hanna > >> > >> -- > >> Department of Plant Anatomy and Cytology Faculty of Biology and > >> Environmental Protection University of Silesia Jagiellonska Str. 28 > >> 40-032 Katowice > >> Poland > >> > > > > > > > > -- > > Tineke Vendrig, ing > > technical engeneer optical microscopy > > Delft University of Technology > > Bionano Science > > Kavli Institute of Nanoscience > > Lorentzweg 1 > > 2628LJ Delft > > room F185 > > Tel: +31 27 89299 > > Fax:+31 15 2781202 > > email: [hidden email] > > mobile phone: 06-34241412 > > > > > > -- > Department of Plant Anatomy and Cytology Faculty of Biology and > Environmental Protection University of Silesia Jagiellonska Str. 28 > 40-032 Katowice > Poland  | 
			
 
	
	 | 
						Julio Vazquez | 
| 
					
		 | 
				
					
	 
				In reply to this post by Hanna_SN
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** When we see similar problems, it tends to be because someone inadvertently bumped a shutter.... start near the light source and work your way along the light path up to the objective, and move any lever or slider you encounter. Our Olympus IX70 has a manual shutter which is a horizontal rotating disk between the filter turret and the objectives. People often bump into the lever and close it partially. On other microscopes, the shutter may be between the lamp and the filter turret. Also, if you have any DIC components, make sure they are either in or out.... sometimes they are half way, and will block part of the illumination field.... -- Julio Vazquez Fred Hutchinson Cancer Research Center Seattle, WA http://www.fhcrc.org/ On Oct 6, 2011, at 7:20 AM, Hanna Sas Nowosielska wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear all, > > For a couple of months we are struggling with a serious problem with > the field illumination in our Olympus Scan^R Scanning station. The > image of the specimen, when observed on the computer screen is much > darker on one side than on the other. We thought it is the lamp but > the lamp is new at properly inserted. Then we tried calibrating the > optic fiber, but it didn't help at all. Our third guess was the > filters.We checked all of the filters - the coating has lots of > small dark spots with biggest concentration of the spots in the middle > of the filters, but I am concerned if that can produce uneven > illumination of the field so that only one part (let say left corner) > is underilluminated? So, I turn to you and your experience: did may > anyone have the similar problem? Do you have any suggestion what can > be the cause of the problem? > > Best regards, > Hanna > > -- > Department of Plant Anatomy and Cytology > Faculty of Biology and Environmental Protection > University of Silesia > Jagiellonska Str. 28 > 40-032 Katowice > Poland  | 
			
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I got a rather left field enquiry today, as to whether there were calibration standards for quantum yield. It seems that the person wants to measure quantum yield under the microscope. My immediate response was that this is impossible. Quantum yield is easy enough to measure in a cuvette but would it be possible in a microscope? You could make a standard of a known concentration of fluorescein in a cell made by a spacer under the coverslip, but where do you go from there, if both quantum yield and extinction coefficient of the test sample are unknown? Any bright ideas? Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net  | 
			
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** If you knew the focal volume being illuminated, you could guess at how many dye molecules were present based on concentration. Then based on input power figure out the photon flux rate through the volume. The tricky part would be figuring out what the chances are of any one dye molecule being struck by a photon. If you had a photon counting detection system you could then compare detected photons vs inputted photons. This would have to be modified by the probability of a photon hitting a dye molecule in the first place and the QE of your detector (so for every detected emitted photon roughly how many do you fail to detected, and of course photon density in the illuminated volume. These are just my first thoughts. I'd have to give it more consideration, especially regarding the likelihood of a photon hitting a dye molecule and how that would modify the final value. Craig On Thu, Oct 6, 2011 at 7:43 PM, Guy Cox <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I got a rather left field enquiry today, as to whether there were > calibration standards for quantum yield. It seems that the person wants > to measure quantum yield under the microscope. My immediate response > was that this is impossible. Quantum yield is easy enough to measure in > a cuvette but would it be possible in a microscope? You could make a > standard of a known concentration of fluorescein in a cell made by a > spacer under the coverslip, but where do you go from there, if both > quantum yield and extinction coefficient of the test sample are unknown? > > Any bright ideas? > > Guy > > > > > Optical Imaging Techniques in Cell Biology > by Guy Cox CRC Press / Taylor & Francis > http://www.guycox.com/optical.htm > ______________________________________________ > Associate Professor Guy Cox, MA, DPhil(Oxon) > Australian Centre for Microscopy & Microanalysis, > Madsen Building F09, University of Sydney, NSW 2006 > > Phone +61 2 9351 3176 Fax +61 2 9351 7682 > Mobile 0413 281 861 > ______________________________________________ > http://www.guycox.net > >  | 
			
| 
					
		 | 
				
					
	 
				In reply to this post by Guy Cox-2
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Is the test sample a biological tissue or a solution of a chemical? -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Guy Cox Sent: Thursday, October 06, 2011 9:44 PM To: [hidden email] Subject: Quantum yield ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I got a rather left field enquiry today, as to whether there were calibration standards for quantum yield. It seems that the person wants to measure quantum yield under the microscope. My immediate response was that this is impossible. Quantum yield is easy enough to measure in a cuvette but would it be possible in a microscope? You could make a standard of a known concentration of fluorescein in a cell made by a spacer under the coverslip, but where do you go from there, if both quantum yield and extinction coefficient of the test sample are unknown? Any bright ideas? Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net  | 
			
| 
					
		 | 
				
					
	 
				In reply to this post by Guy Cox-2
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** In order to measure quantum yields by the reference method, one only needs a defined measurement volume, a standard dye of known absorbance (at the microscope excitation wavelength) and quantum yield, and the absorbance of your sample at the microscope excitation wavelength. Quantum yield is then defined as QY=QYstandard*(A/Astandard) *(Fstandard/F)*(n^2/nstandard^2). Here n is the refractive index, F is the measured fluorescence intensity and A is the absorbance at the microscope excitation wavelength. Note that it is important that the standard and your unknown have similar emission spectra. Otherwise you will have to correct for the wavelength dependence of your microscope detection efficiency. Of course, if you have enough sample to measure absorbance, you can also measure fluorescence in a cuvette and then there is no point in going to the microscope. Jay -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Guy Cox Sent: Thursday, October 06, 2011 8:44 PM To: [hidden email] Subject: Quantum yield ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I got a rather left field enquiry today, as to whether there were calibration standards for quantum yield. It seems that the person wants to measure quantum yield under the microscope. My immediate response was that this is impossible. Quantum yield is easy enough to measure in a cuvette but would it be possible in a microscope? You could make a standard of a known concentration of fluorescein in a cell made by a spacer under the coverslip, but where do you go from there, if both quantum yield and extinction coefficient of the test sample are unknown? Any bright ideas? Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net  | 
			
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** That's a pretty neat method! It avoids having to worry about figuring out photon flux by calibrating against a known. Where would you get trustworthy QE calibration standards though? Craig On Fri, Oct 7, 2011 at 10:14 AM, Unruh, Jay <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > In order to measure quantum yields by the reference method, one only needs > a defined measurement volume, a standard dye of known absorbance (at the > microscope excitation wavelength) and quantum yield, and the absorbance of > your sample at the microscope excitation wavelength. Quantum yield is then > defined as QY=QYstandard*(A/Astandard) *(Fstandard/F)*(n^2/nstandard^2). > Here n is the refractive index, F is the measured fluorescence intensity > and A is the absorbance at the microscope excitation wavelength. Note that > it is important that the standard and your unknown have similar emission > spectra. Otherwise you will have to correct for the wavelength dependence > of your microscope detection efficiency. Of course, if you have enough > sample to measure absorbance, you can also measure fluorescence in a cuvette > and then there is no point in going to the microscope. > > Jay > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Guy Cox > Sent: Thursday, October 06, 2011 8:44 PM > To: [hidden email] > Subject: Quantum yield > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I got a rather left field enquiry today, as to whether there were > calibration standards for quantum yield. It seems that the person wants to > measure quantum yield under the microscope. My immediate response was that > this is impossible. Quantum yield is easy enough to measure in a cuvette > but would it be possible in a microscope? You could make a standard of a > known concentration of fluorescein in a cell made by a > spacer under the coverslip, but where do you go from there, if both > quantum yield and extinction coefficient of the test sample are unknown? > > Any bright ideas? > > Guy > > > > > Optical Imaging Techniques in Cell Biology > by Guy Cox CRC Press / Taylor & Francis > http://www.guycox.com/optical.htm > ______________________________________________ > Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for > Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW > 2006 > > Phone +61 2 9351 3176 Fax +61 2 9351 7682 > Mobile 0413 281 861 > ______________________________________________ > http://www.guycox.net > >  | 
			
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** We used to create a defined measurement volume by placing a liquid sample between a slide and a spherical lens - then you get a range of depths that increase with the distance from the center. Cytometry 75A, 874-881 (2009) Mike -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau Sent: Friday, October 07, 2011 12:53 PM To: [hidden email] Subject: Re: Quantum yield On Fri, Oct 7, 2011 at 10:14 AM, Unruh, Jay <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > In order to measure quantum yields by the reference method, one only needs > a defined measurement volume, a standard dye of known absorbance (at the > microscope excitation wavelength) and quantum yield, and the absorbance of > your sample at the microscope excitation wavelength. Quantum yield is then > defined as QY=QYstandard*(A/Astandard) *(Fstandard/F)*(n^2/nstandard^2). > Here n is the refractive index, F is the measured fluorescence intensity > and A is the absorbance at the microscope excitation wavelength. Note that > it is important that the standard and your unknown have similar emission > spectra. Otherwise you will have to correct for the wavelength dependence > of your microscope detection efficiency. Of course, if you have enough > sample to measure absorbance, you can also measure fluorescence in a cuvette > and then there is no point in going to the microscope. > > Jay > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Guy Cox > Sent: Thursday, October 06, 2011 8:44 PM > To: [hidden email] > Subject: Quantum yield > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I got a rather left field enquiry today, as to whether there were > calibration standards for quantum yield. It seems that the person wants to > measure quantum yield under the microscope. My immediate response was that > this is impossible. Quantum yield is easy enough to measure in a cuvette > but would it be possible in a microscope? You could make a standard of a > known concentration of fluorescein in a cell made by a > spacer under the coverslip, but where do you go from there, if both > quantum yield and extinction coefficient of the test sample are unknown? > > Any bright ideas? > > Guy > > > > > Optical Imaging Techniques in Cell Biology > by Guy Cox CRC Press / Taylor & Francis > http://www.guycox.com/optical.htm > ______________________________________________ > Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for > Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW > 2006 > > Phone +61 2 9351 3176 Fax +61 2 9351 7682 > Mobile 0413 281 861 > ______________________________________________ > http://www.guycox.net > >  | 
			
| 
					
		 | 
				
					
	 
				In reply to this post by Craig Brideau
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Oops--I accidentally inverted the equation. It should be QY=QYstandard*(F/Fstandard) *(Astandard/A)*(n^2/nstandard^2). For quantum yield standards, see Magde et al photochemistry and photobiology 2002 vol 75 p 327. It gives a thorough study for fluorescein and rhodamine 6G. Other wavelengths are a bit more difficult. Lakowicz principles of fluorescence spectroscopy gives more details and a few standards from the literature. Jay -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Craig Brideau Sent: Friday, October 07, 2011 11:53 AM To: [hidden email] Subject: Re: Quantum yield ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** That's a pretty neat method! It avoids having to worry about figuring out photon flux by calibrating against a known. Where would you get trustworthy QE calibration standards though? Craig On Fri, Oct 7, 2011 at 10:14 AM, Unruh, Jay <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > In order to measure quantum yields by the reference method, one only > needs a defined measurement volume, a standard dye of known absorbance > (at the microscope excitation wavelength) and quantum yield, and the > absorbance of your sample at the microscope excitation wavelength. > Quantum yield is then defined as QY=QYstandard*(A/Astandard) *(Fstandard/F)*(n^2/nstandard^2). > Here n is the refractive index, F is the measured fluorescence > intensity and A is the absorbance at the microscope excitation > wavelength. Note that it is important that the standard and your > unknown have similar emission spectra. Otherwise you will have to > correct for the wavelength dependence of your microscope detection > efficiency. Of course, if you have enough sample to measure > absorbance, you can also measure fluorescence in a cuvette and then there is no point in going to the microscope. > > Jay > > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] > On Behalf Of Guy Cox > Sent: Thursday, October 06, 2011 8:44 PM > To: [hidden email] > Subject: Quantum yield > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I got a rather left field enquiry today, as to whether there were > calibration standards for quantum yield. It seems that the person > wants to measure quantum yield under the microscope. My immediate > response was that this is impossible. Quantum yield is easy enough to > measure in a cuvette but would it be possible in a microscope? You > could make a standard of a known concentration of fluorescein in a cell made by a > spacer under the coverslip, but where do you go from there, if both > quantum yield and extinction coefficient of the test sample are unknown? > > Any bright ideas? > > Guy > > > > > Optical Imaging Techniques in Cell Biology > by Guy Cox CRC Press / Taylor & Francis > http://www.guycox.com/optical.htm > ______________________________________________ > Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for > Microscopy & Microanalysis, Madsen Building F09, University of Sydney, > NSW > 2006 > > Phone +61 2 9351 3176 Fax +61 2 9351 7682 > Mobile 0413 281 861 > ______________________________________________ > http://www.guycox.net > >  | 
			
| 
					
		 | 
				
					
	 
				In reply to this post by mmodel
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** It's quantum dots, just to compound the problem. Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of MODEL, MICHAEL Sent: Saturday, 8 October 2011 12:19 AM To: [hidden email] Subject: Re: Quantum yield ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Is the test sample a biological tissue or a solution of a chemical? -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Guy Cox Sent: Thursday, October 06, 2011 9:44 PM To: [hidden email] Subject: Quantum yield ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I got a rather left field enquiry today, as to whether there were calibration standards for quantum yield. It seems that the person wants to measure quantum yield under the microscope. My immediate response was that this is impossible. Quantum yield is easy enough to measure in a cuvette but would it be possible in a microscope? You could make a standard of a known concentration of fluorescein in a cell made by a spacer under the coverslip, but where do you go from there, if both quantum yield and extinction coefficient of the test sample are unknown? Any bright ideas? Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net ----- No virus found in this message. Checked by AVG - www.avg.com Version: 10.0.1410 / Virus Database: 1520/3942 - Release Date: 10/06/11  | 
			
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** If the extinction coefficient of their suspension is known at a wavelength where some well-characterized fluorophore also absorbs, then their fluorescence intensities could be compared under identical conditions... But then, there is an issue of detector's spectral sensitivity, and even for some well-charactrized fluorophores, different people report very different values of quantum yield, I don't know why. I hope others would have something more helpful to say... Mike ________________________________________ From: Confocal Microscopy List [[hidden email]] On Behalf Of Guy Cox [[hidden email]] Sent: Saturday, October 08, 2011 2:21 AM To: [hidden email] Subject: Re: Quantum yield ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** It's quantum dots, just to compound the problem. Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of MODEL, MICHAEL Sent: Saturday, 8 October 2011 12:19 AM To: [hidden email] Subject: Re: Quantum yield ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Is the test sample a biological tissue or a solution of a chemical? -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Guy Cox Sent: Thursday, October 06, 2011 9:44 PM To: [hidden email] Subject: Quantum yield ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I got a rather left field enquiry today, as to whether there were calibration standards for quantum yield. It seems that the person wants to measure quantum yield under the microscope. My immediate response was that this is impossible. Quantum yield is easy enough to measure in a cuvette but would it be possible in a microscope? You could make a standard of a known concentration of fluorescein in a cell made by a spacer under the coverslip, but where do you go from there, if both quantum yield and extinction coefficient of the test sample are unknown? Any bright ideas? Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net ----- No virus found in this message. Checked by AVG - www.avg.com Version: 10.0.1410 / Virus Database: 1520/3942 - Release Date: 10/06/11  | 
			
| 
					
		 | 
				
					
	 
				In reply to this post by Guy Cox-2
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Any of the conventional macroscopic methods of quantum yield determination will be susceptible to large errors if implemented under a microscope as the optical pathlength is so short. Even in a cuvette, the precision of the macroscopic methods is not so good (generally about +/- 5%). Particularly in the case of quantum dots, single molecule approaches make more sense and are more physically informative than ensemble measurements: http://www.ncbi.nlm.nih.gov/sites/entrez?cmd=Retrieve&db=PubMed&dopt=AbstractPlus&list_uids=16169907 Iain Iain Johnson Consulting Eugene, OR On Fri, Oct 7, 2011 at 11:21 PM, Guy Cox <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > It's quantum dots, just to compound the problem. > > Guy > > Optical Imaging Techniques in Cell Biology > by Guy Cox CRC Press / Taylor & Francis > http://www.guycox.com/optical.htm > ______________________________________________ > Associate Professor Guy Cox, MA, DPhil(Oxon) > Australian Centre for Microscopy & Microanalysis, > Madsen Building F09, University of Sydney, NSW 2006 > > Phone +61 2 9351 3176 Fax +61 2 9351 7682 > Mobile 0413 281 861 > ______________________________________________ > http://www.guycox.net > > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of MODEL, MICHAEL > Sent: Saturday, 8 October 2011 12:19 AM > To: [hidden email] > Subject: Re: Quantum yield > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Is the test sample a biological tissue or a solution of a chemical? > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Guy Cox > Sent: Thursday, October 06, 2011 9:44 PM > To: [hidden email] > Subject: Quantum yield > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I got a rather left field enquiry today, as to whether there were > calibration standards for quantum yield. It seems that the person wants > to measure quantum yield under the microscope. My immediate response > was that this is impossible. Quantum yield is easy enough to measure in > a cuvette but would it be possible in a microscope? You could make a > standard of a known concentration of fluorescein in a cell made by a > spacer under the coverslip, but where do you go from there, if both > quantum yield and extinction coefficient of the test sample are unknown? > > Any bright ideas? > > Guy > > > > > Optical Imaging Techniques in Cell Biology > by Guy Cox CRC Press / Taylor & Francis > http://www.guycox.com/optical.htm > ______________________________________________ > Associate Professor Guy Cox, MA, DPhil(Oxon) > Australian Centre for Microscopy & Microanalysis, > Madsen Building F09, University of Sydney, NSW 2006 > > Phone +61 2 9351 3176 Fax +61 2 9351 7682 > Mobile 0413 281 861 > ______________________________________________ > http://www.guycox.net > > > ----- > No virus found in this message. > Checked by AVG - www.avg.com > Version: 10.0.1410 / Virus Database: 1520/3942 - Release Date: 10/06/11 >  | 
			
 
	
	 | 
						Axel Kurt Preuss | 
| 
					
		 | 
				
					
	 
				In reply to this post by Hanna_SN
			 
	
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Hanna Koehler comes to mind as already suggested. I m not familiar with scanning station We had similar problem on an Olympus scope and found that the glass in front of the transmission lamp had a crack (heat?) So, this glass in the transmission housing, which directly faces the microscope stage , had a crack and that produced a shadow in the image pretending to be a Koehler problem cheers Axel 6-19B , 92715622 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Hanna Sas Nowosielska Sent: Thursday, 6 October, 2011 10:37 PM To: [hidden email] Subject: Re: Rejected posting to [hidden email] ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Yes, at least Olymus claims everything with the optical path is ok. Hanna 2011/10/6 tineke vendrig <[hidden email]>: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Have you optimised you microscope ?? Kohler illumination and so on..?? > > Tineke Vendrig, ing > technical engeneer optical microscopy > Delft University of Technology > Bionano Science > Kavli Institute of Nanoscience > Lorentzweg 1 > 2628LJ Delft > room F185 > Tel: +31 27 89299 > Fax:+31 15 2781202 > email: [hidden email] > mobile phone: 06-34241412 > > > 2011/10/6 Hanna Sas Nowosielska <[hidden email]> > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Dear all, >> >> For a couple of months we are struggling with a serious problem with >> the field illumination in our Olympus Scan^R Scanning station. The >> image of the specimen, when observed on the computer screen is much >> darker on one side than on the other. We thought it is the lamp but >> the lamp is new at properly inserted. Then we tried calibrating the >> optic fiber, but it didn't help at all. Our third guess was the >> filters.We checked all of the filters - the coating has lots of >> small dark spots with biggest concentration of the spots in the >> middle of the filters, but I am concerned if that can produce uneven >> illumination of the field so that only one part (let say left corner) >> is underilluminated? So, I turn to you and your experience: did may >> anyone have the similar problem? Do you have any suggestion what can >> be the cause of the problem? >> >> Best regards, >> Hanna >> >> -- >> Department of Plant Anatomy and Cytology Faculty of Biology and >> Environmental Protection University of Silesia Jagiellonska Str. 28 >> 40-032 Katowice >> Poland >> > > > > -- > Tineke Vendrig, ing > technical engeneer optical microscopy > Delft University of Technology > Bionano Science > Kavli Institute of Nanoscience > Lorentzweg 1 > 2628LJ Delft > room F185 > Tel: +31 27 89299 > Fax:+31 15 2781202 > email: [hidden email] > mobile phone: 06-34241412 > -- Department of Plant Anatomy and Cytology Faculty of Biology and Environmental Protection University of Silesia Jagiellonska Str. 28 40-032 Katowice Poland Note: This message may contain confidential information. If this Email/Fax has been sent to you by mistake, please notify the sender and delete it immediately. Thank you.  | 
			
 
	
	 | 
						Zac Arrac Atelaz | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I was thinking on this one, the uneveness of the field is more visible in a specific wavelength or in all of them? usually UV is the most uneven. A different case might be the DIC ilumination being in the wrong position (usually it has 2 positions), the other one would be a crack in the fiber (hopefully this one is just a wild guess) Best luck Gabriel OH > Date: Thu, 3 Nov 2011 11:46:58 +0800 > From: [hidden email] > Subject: Re: Rejected posting to [hidden email] > To: [hidden email] > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear Hanna > Koehler comes to mind as already suggested. > I m not familiar with scanning station > > We had similar problem on an Olympus scope and found that the glass in front of the transmission lamp had a crack (heat?) > > So, this glass in the transmission housing, which directly faces the microscope stage , had a crack and that produced a shadow in the image pretending to be a Koehler problem > > cheers > > Axel 6-19B , 92715622 > > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Hanna Sas Nowosielska > Sent: Thursday, 6 October, 2011 10:37 PM > To: [hidden email] > Subject: Re: Rejected posting to [hidden email] > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Yes, at least Olymus claims everything with the optical path is ok. > > Hanna > > 2011/10/6 tineke vendrig <[hidden email]>: > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Have you optimised you microscope ?? Kohler illumination and so on..?? > > > > Tineke Vendrig, ing > > technical engeneer optical microscopy > > Delft University of Technology > > Bionano Science > > Kavli Institute of Nanoscience > > Lorentzweg 1 > > 2628LJ Delft > > room F185 > > Tel: +31 27 89299 > > Fax:+31 15 2781202 > > email: [hidden email] > > mobile phone: 06-34241412 > > > > > > 2011/10/6 Hanna Sas Nowosielska <[hidden email]> > > > >> ***** > >> To join, leave or search the confocal microscopy listserv, go to: > >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >> ***** > >> > >> Dear all, > >> > >> For a couple of months we are struggling with a serious problem with > >> the field illumination in our Olympus Scan^R Scanning station. The > >> image of the specimen, when observed on the computer screen is much > >> darker on one side than on the other. We thought it is the lamp but > >> the lamp is new at properly inserted. Then we tried calibrating the > >> optic fiber, but it didn't help at all. Our third guess was the > >> filters.We checked all of the filters - the coating has lots of > >> small dark spots with biggest concentration of the spots in the > >> middle of the filters, but I am concerned if that can produce uneven > >> illumination of the field so that only one part (let say left corner) > >> is underilluminated? So, I turn to you and your experience: did may > >> anyone have the similar problem? Do you have any suggestion what can > >> be the cause of the problem? > >> > >> Best regards, > >> Hanna > >> > >> -- > >> Department of Plant Anatomy and Cytology Faculty of Biology and > >> Environmental Protection University of Silesia Jagiellonska Str. 28 > >> 40-032 Katowice > >> Poland > >> > > > > > > > > -- > > Tineke Vendrig, ing > > technical engeneer optical microscopy > > Delft University of Technology > > Bionano Science > > Kavli Institute of Nanoscience > > Lorentzweg 1 > > 2628LJ Delft > > room F185 > > Tel: +31 27 89299 > > Fax:+31 15 2781202 > > email: [hidden email] > > mobile phone: 06-34241412 > > > > > > -- > Department of Plant Anatomy and Cytology Faculty of Biology and Environmental Protection University of Silesia Jagiellonska Str. 28 > 40-032 Katowice > Poland > > Note: This message may contain confidential information. If this Email/Fax has been sent to you by mistake, please notify the sender and delete it immediately. Thank you.  | 
			
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I was doing a bit of historical research today and I looked up Jan Wichmann, of the historic first paper on STED (Hell & Wichmann, 1994, Optics Letters 19: 780-782). To my surprise the Scopus database showed just one paper for Wichmann - that one, with more than 500 citations. This must be a pretty rare situation. Does anyone know who Wichmann (from Turku in Finland) was and what became of him? Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net  | 
			
 
	
	 | 
						Sudipta Maiti | 
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** According to Prof. Jurgen Wolfrum, Wichmann's Ph.D. advisor and currently a visitor here, Wichmann left science without finishing his Ph.D. for personal reasons. He had taken a leave for a few months during his Ph.D. period to pursue the STED idea with Stefan Hell. Sudipta On Mon, 7 Nov 2011 21:49:36 +1100, Guy Cox wrote > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I was doing a bit of historical research today and I looked up Jan > Wichmann, of the historic first paper on STED (Hell & Wichmann, 1994, > Optics Letters 19: 780-782). To my surprise the Scopus database showed > just one paper for Wichmann - that one, with more than 500 citations. > This must be a pretty rare situation. Does anyone know who Wichmann > (from Turku in Finland) was and what became of him? > > Guy > > Optical Imaging Techniques in Cell Biology > by Guy Cox CRC Press / Taylor & Francis > http://www.guycox.com/optical.htm > ______________________________________________ > Associate Professor Guy Cox, MA, DPhil(Oxon) > Australian Centre for Microscopy & Microanalysis, > Madsen Building F09, University of Sydney, NSW 2006 > > Phone +61 2 9351 3176 Fax +61 2 9351 7682 > Mobile 0413 281 861 > ______________________________________________ > http://www.guycox.net Dr. Sudipta Maiti Dept. of Chemical Sciences Tata Institute of Fundamental Research Homi Bhabha Raod, Colaba, Mumbai 400005 Ph. 91-22-2278-2716 / 2539 Fax: 91-22-2280-4610 alternate e-mail: [hidden email] url: biophotonics.weebly.com  | 
			
| 
					
		 | 
				
					
	
	 
		*****
 
				To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Sudipta, Thanks for this - it is very interesting. You will have noticed that Vitaly Boyko dug out a Jan Wichmann in Finland, and I have contacted him. He was not the Jan Wichmann of the Wichmann and Hell paper. So does Jurgen know what the 'real' Jan Wichmann ended up doing? Guy Optical Imaging Techniques in Cell Biology by Guy Cox CRC Press / Taylor & Francis http://www.guycox.com/optical.htm ______________________________________________ Associate Professor Guy Cox, MA, DPhil(Oxon) Australian Centre for Microscopy & Microanalysis, Madsen Building F09, University of Sydney, NSW 2006 Phone +61 2 9351 3176 Fax +61 2 9351 7682 Mobile 0413 281 861 ______________________________________________ http://www.guycox.net -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Sudipta Maiti Sent: Saturday, 12 November 2011 9:43 PM To: [hidden email] Subject: Re: Historical question - Jan Wichmann ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** According to Prof. Jurgen Wolfrum, Wichmann's Ph.D. advisor and currently a visitor here, Wichmann left science without finishing his Ph.D. for personal reasons. He had taken a leave for a few months during his Ph.D. period to pursue the STED idea with Stefan Hell. Sudipta On Mon, 7 Nov 2011 21:49:36 +1100, Guy Cox wrote > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > I was doing a bit of historical research today and I looked up Jan > Wichmann, of the historic first paper on STED (Hell & Wichmann, 1994, > Optics Letters 19: 780-782). To my surprise the Scopus database showed > just one paper for Wichmann - that one, with more than 500 citations. > This must be a pretty rare situation. Does anyone know who Wichmann > (from Turku in Finland) was and what became of him? > > Guy > > Optical Imaging Techniques in Cell Biology > by Guy Cox CRC Press / Taylor & Francis > http://www.guycox.com/optical.htm > ______________________________________________ > Associate Professor Guy Cox, MA, DPhil(Oxon) > Australian Centre for Microscopy & Microanalysis, > Madsen Building F09, University of Sydney, NSW 2006 > > Phone +61 2 9351 3176 Fax +61 2 9351 7682 > Mobile 0413 281 861 > ______________________________________________ > http://www.guycox.net Dr. Sudipta Maiti Dept. of Chemical Sciences Tata Institute of Fundamental Research Homi Bhabha Raod, Colaba, Mumbai 400005 Ph. 91-22-2278-2716 / 2539 Fax: 91-22-2280-4610 alternate e-mail: [hidden email] url: biophotonics.weebly.com ----- No virus found in this message. Checked by AVG - www.avg.com Version: 10.0.1411 / Virus Database: 2092/4011 - Release Date: 11/11/11  | 
			