Replacing diode in 405 nm laser

classic Classic list List threaded Threaded
♦
♦
3 messages Options
Angela V Klaus Angela V Klaus
Reply | Threaded
Open this post in threaded view
♦
♦
|

Replacing diode in 405 nm laser

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Hi Listers,

Our 405 nm laser on our confocal might be burned out (haven't had a chance to troubleshoot yet).  I have a quote for replacing the laser head, but was wondering if it would be possible for me to just replace the diode myself.  There seem to be many sources for 405 nm diodes.

Many thanks for any insight.

Best,

Angela

Angela V. Klaus, PhD
Assistant Professor
Department of Biological Sciences
McNulty Science and Technology Center, Rm. 315
Seton Hall University
400 South Orange Ave.
South Orange, NJ 07079
(v): 973-761-9146
(fax): 973-275-2905
(email): [hidden email]
Steve Barlow Steve Barlow
Reply | Threaded
Open this post in threaded view
♦
♦
|

how are you?

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Angela

So nice to see your name in print.  How are you doing these days?  How is life as Assistant Professor?  Do you miss NSF?  Last year we snagged one of the special MRI grants--the one time of late no match was required-- for a new FEI FEG Quanta 450 SEM with Oxford x-ray, hot and cold stages, CL detector, new coater etc.  What a difference between my old analog hitachi and this new computer wonder.  Course, it costs me nothing to maintain my 15 year old hitachi vs the 35k of the new wunderkid.

And of course nobody in the CalState system knows if we will have jobs from one month to the next, as the legislature plays with smoke and mirror budgets and no new taxes yahoos.

Hope this note finds you well.

Steve
On Jun 2, 2011, at 5:51 AM, Angela V Klaus wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Listers,
>
> Our 405 nm laser on our confocal might be burned out (haven't had a chance to troubleshoot yet).  I have a quote for replacing the laser head, but was wondering if it would be possible for me to just replace the diode myself.  There seem to be many sources for 405 nm diodes.
>
> Many thanks for any insight.
>
> Best,
>
> Angela
>
> Angela V. Klaus, PhD
> Assistant Professor
> Department of Biological Sciences
> McNulty Science and Technology Center, Rm. 315
> Seton Hall University
> 400 South Orange Ave.
> South Orange, NJ 07079
> (v): 973-761-9146
> (fax): 973-275-2905
> (email): [hidden email]
>
>

Steve Barlow, Ph.D.
EM Facility/SDSU Biology
5500 Campanile Drive
MC-4614
San Diego, CA. 92182-4614
Phone: (619)594-4523
Fax: (619)594-5676
email: [hidden email]
web: www.sci.sdsu.edu/emfacility
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Replacing diode in 405 nm laser

In reply to this post by Angela V Klaus
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****

Different diodes have different divergences, usually astigmatic, so the
optics have to be customized for the diode.  If you had an exact replacement
for the defunct diode you could replace it, but the laser would still have
to be realigned internally.  Basically the cost of the laser is wrapped up
in the man-hours to align it and the like, so you probably wouldn't save
much money trying to do this.  Finally, most of the cost of a 405 laser *is*
the cost of the diode.  They go for anywhere from $300 to $1500 dollars
depending on the output power.

Craig



On Thu, Jun 2, 2011 at 6:51 AM, Angela V Klaus <[hidden email]> wrote:

> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi Listers,
>
> Our 405 nm laser on our confocal might be burned out (haven't had a chance
> to troubleshoot yet).  I have a quote for replacing the laser head, but was
> wondering if it would be possible for me to just replace the diode myself.
>  There seem to be many sources for 405 nm diodes.
>
> Many thanks for any insight.
>
> Best,
>
> Angela
>
> Angela V. Klaus, PhD
> Assistant Professor
> Department of Biological Sciences
> McNulty Science and Technology Center, Rm. 315
> Seton Hall University
> 400 South Orange Ave.
> South Orange, NJ 07079
> (v): 973-761-9146
> (fax): 973-275-2905
> (email): [hidden email]
>