To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Confocalists, We’re trying to circumvent the annoyance of resonant mirrors screaming at 8 kHz. I was wondering whether anyone has a recommendation for constructing soundproof containment? Generally were using inexpensive resonant mirrors from EOPC: http://www.eopc.com/sc30.html Also, does anyone know whether the more costly resonant scanners used for confocal/multi-photon laser scanning microscopies are any better in this regard? Were not hugely concerned with other attributes (simply for light sheet microscopy stripe suppression) just the piercing noise. Many thanks in advance! Rory ………………………………………………………… Dr. Rory Power Morgridge Institute for Research Medical Engineering | Huisken Lab 330 N Orchard St, Madison, WI, 53715 Tel: 608 316 4554 [hidden email]<mailto:[hidden email]> |
![]() ![]() |
Michael Giacomelli |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Rory, 8 KHz unidirectional or bidirectional (so 4 KHz fundamental)? The 8 KHz scanners are usually not that bad even at maximum amplitude because the 8 kHz tone is a lot harder to hear and essentially all the harmonics are out of audible range. For 4 KHz at larger deflection angles you either need hearing protection or to really isolate it from the room. A metal box around the scanner and a window on the input port so that air cannot pass in/out of the scanner helps a lot. This is how most commerical systems work. I don't think the model matters; the noise comes from pushing air, so for a given mirror size and deflection angle the noise power is going to be almost constant. Mike On Wed, Jul 25, 2018 at 9:03 PM, Power, Rory <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Confocalists, > > We’re trying to circumvent the annoyance of resonant mirrors screaming at > 8 kHz. I was wondering whether anyone has a recommendation for constructing > soundproof containment? > > Generally were using inexpensive resonant mirrors from EOPC: > http://www.eopc.com/sc30.html > > Also, does anyone know whether the more costly resonant scanners used for > confocal/multi-photon laser scanning microscopies are any better in this > regard? Were not hugely concerned with other attributes (simply for light > sheet microscopy stripe suppression) just the piercing noise. > > Many thanks in advance! > > Rory > > ………………………………………………………… > > Dr. Rory Power > Morgridge Institute for Research > Medical Engineering | Huisken Lab > 330 N Orchard St, Madison, WI, 53715 > Tel: 608 316 4554 > [hidden email]<mailto:[hidden email]> > > > > > ... [show rest of quote]
|
![]() ![]() |
Power, Rory |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Mike, Thanks for your reply. I was rather referring to an 8 kHz fundamental oscillation. The noise is quite unpleasant - I’d hate to hear the 4 kHz if that’s worse as you say. Isolation should be possible. Perhaps combined with some foam we’ll arrive at a workable system. Yes, I assumed the same. Thanks again. Regards Rory ………………………………………………………… Dr. Rory Power HFSP Postdoctoral Fellow Morgridge Institute for Research Medical Engineering | Huisken Lab 330 N Orchard St, Madison, WI, 53715 Tel: 608 316 4554 [hidden email]<mailto:[hidden email]> On Jul 25, 2018, at 21:18, Michael Giacomelli <[hidden email]<mailto:[hidden email]>> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Rory, 8 KHz unidirectional or bidirectional (so 4 KHz fundamental)? The 8 KHz scanners are usually not that bad even at maximum amplitude because the 8 kHz tone is a lot harder to hear and essentially all the harmonics are out of audible range. For 4 KHz at larger deflection angles you either need hearing protection or to really isolate it from the room. A metal box around the scanner and a window on the input port so that air cannot pass in/out of the scanner helps a lot. This is how most commerical systems work. I don't think the model matters; the noise comes from pushing air, so for a given mirror size and deflection angle the noise power is going to be almost constant. Mike On Wed, Jul 25, 2018 at 9:03 PM, Power, Rory <[hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Confocalists, We’re trying to circumvent the annoyance of resonant mirrors screaming at 8 kHz. I was wondering whether anyone has a recommendation for constructing soundproof containment? Generally were using inexpensive resonant mirrors from EOPC: http://www.eopc.com/sc30.html Also, does anyone know whether the more costly resonant scanners used for confocal/multi-photon laser scanning microscopies are any better in this regard? Were not hugely concerned with other attributes (simply for light sheet microscopy stripe suppression) just the piercing noise. Many thanks in advance! Rory ………………………………………………………… Dr. Rory Power Morgridge Institute for Research Medical Engineering | Huisken Lab 330 N Orchard St, Madison, WI, 53715 Tel: 608 316 4554 [hidden email]<mailto:[hidden email]> |
![]() ![]() |
Christian Feldhaus |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi, for getting rid of some noise in our microscope room I played around with Basotect foam (no commercial interest, other dedicated sound dampening foam will probably do the same job). I found that 20mm Basotect was quite efficient at 8 kHz, reducing the noise level from 11dBA down to 6dBA by just covering half of one of the room walls. For lower frequencies you will need thicker foam. Best, Christian On 26.07.2018 04:33, Power, Rory wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Mike, > > Thanks for your reply. I was rather referring to an 8 kHz fundamental oscillation. The noise is quite unpleasant - I’d hate to hear the 4 kHz if that’s worse as you say. > > Isolation should be possible. Perhaps combined with some foam we’ll arrive at a workable system. > > Yes, I assumed the same. Thanks again. > > Regards > > Rory > > ………………………………………………………… > > Dr. Rory Power > HFSP Postdoctoral Fellow > Morgridge Institute for Research > Medical Engineering | Huisken Lab > 330 N Orchard St, Madison, WI, 53715 > Tel: 608 316 4554 > [hidden email]<mailto:[hidden email]> > > > > > On Jul 25, 2018, at 21:18, Michael Giacomelli <[hidden email]<mailto:[hidden email]>> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Rory, > > 8 KHz unidirectional or bidirectional (so 4 KHz fundamental)? The 8 KHz > scanners are usually not that bad even at maximum amplitude because the 8 > kHz tone is a lot harder to hear and essentially all the harmonics are out > of audible range. > > For 4 KHz at larger deflection angles you either need hearing protection or > to really isolate it from the room. A metal box around the scanner and a > window on the input port so that air cannot pass in/out of the scanner > helps a lot. This is how most commerical systems work. > > I don't think the model matters; the noise comes from pushing air, so for a > given mirror size and deflection angle the noise power is going to be > almost constant. > > Mike > > On Wed, Jul 25, 2018 at 9:03 PM, Power, Rory <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Confocalists, > > We’re trying to circumvent the annoyance of resonant mirrors screaming at > 8 kHz. I was wondering whether anyone has a recommendation for constructing > soundproof containment? > > Generally were using inexpensive resonant mirrors from EOPC: > http://www.eopc.com/sc30.html > > Also, does anyone know whether the more costly resonant scanners used for > confocal/multi-photon laser scanning microscopies are any better in this > regard? Were not hugely concerned with other attributes (simply for light > sheet microscopy stripe suppression) just the piercing noise. > > Many thanks in advance! > > Rory > > ………………………………………………………… > > Dr. Rory Power > Morgridge Institute for Research > Medical Engineering | Huisken Lab > 330 N Orchard St, Madison, WI, 53715 > Tel: 608 316 4554 > [hidden email]<mailto:[hidden email]> > > > > > > ... [show rest of quote] -- Christian Feldhaus, PhD Light Microscopy Facility MPI for Developmental Biology Max-Planck-Ring 5 (Spemannstrasse 35) 72076 Tübingen Germany deliveries to: Max-Planck-Ring 1 72076 Tübingen Tel.: ++49 7071 601443 |
![]() ![]() |
Mark Cannell-2 |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Power, Rory
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Should be amenable to sound cancellation -small speaker and microphone near scanner with a negative feedback amplifier... HTH Mark B. Cannell. Ph.D. FRSNZ FISHR Department of Physiology, Pharmacology & Neuroscience School of Medical Sciences University Walk Bristol BS8 1TD [hidden email] On 26/07/18, 2:19 AM, "Confocal Microscopy List on behalf of Power, Rory" <[hidden email] on behalf of [hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Confocalists, We’re trying to circumvent the annoyance of resonant mirrors screaming at 8 kHz. I was wondering whether anyone has a recommendation for constructing soundproof containment? Generally were using inexpensive resonant mirrors from EOPC: http://www.eopc.com/sc30.html Also, does anyone know whether the more costly resonant scanners used for confocal/multi-photon laser scanning microscopies are any better in this regard? Were not hugely concerned with other attributes (simply for light sheet microscopy stripe suppression) just the piercing noise. Many thanks in advance! Rory ………………………………………………………… Dr. Rory Power Morgridge Institute for Research Medical Engineering | Huisken Lab 330 N Orchard St, Madison, WI, 53715 Tel: 608 316 4554 [hidden email]<mailto:[hidden email]> |
![]() ![]() |
David C |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Power, Rory
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Rory, I use an 8 kHz scanner from Cambridge Technology and it is loud to the point that I have to wear earplugs in the lab if I'm around for long times. In a commercial confocal/2p intravital microscope, the noise is barely noticeable, which implies that a soundproof container is certainly possible. I don't have any experience with soundproofing, but I would imagine that the scanner could be placed in a completely sealed, custom-made, aluminum enclosure with entry and exit AR-coated windows. best, -- Ph.D. David Chen Postdoc - Myers Lab Max Planck Institute for Molecular Cell Biology and Genetics |
![]() ![]() |
Eric Girard |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Power, Rory
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Rory, Often scanners are enclosed in a vacuum to reduce noise, it may be more complicated to implement than insulation but it would be the most compact approach. Best regards, Eric Girard Sales Engineering / Ingénierie Ventes GiGa Concept Inc Montreal, Quebec, Canada Cell: 514-826-1181 Tel: 514-745-8877 [hidden email] www.gigaconcept.com Meet us at: Photonics West 2018 Booth 4437 OPIE'18 Yokohama CLEO 2018 San José Le mer. 25 juill. 2018 21 h 03, Power, Rory <[hidden email]> a écrit : > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Confocalists, > > We’re trying to circumvent the annoyance of resonant mirrors screaming at > 8 kHz. I was wondering whether anyone has a recommendation for constructing > soundproof containment? > > Generally were using inexpensive resonant mirrors from EOPC: > http://www.eopc.com/sc30.html > > Also, does anyone know whether the more costly resonant scanners used for > confocal/multi-photon laser scanning microscopies are any better in this > regard? Were not hugely concerned with other attributes (simply for light > sheet microscopy stripe suppression) just the piercing noise. > > Many thanks in advance! > > Rory > > ………………………………………………………… > > Dr. Rory Power > Morgridge Institute for Research > Medical Engineering | Huisken Lab > 330 N Orchard St, Madison, WI, 53715 > Tel: 608 316 4554 > [hidden email]<mailto:[hidden email]> > > > > > ... [show rest of quote]
|
![]() ![]() |
Johannes Helm |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Power, Rory
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Good afternoon, Rory: How large is the wavelength band of the laser beam you try to sweep using the scanner? If the wavelength band is not so large (not exceeding 50nm to 100nm), you might get completely rid of the noise by replacing the scanner with an acousto optic deflector, AOD. However, these AODs are chromatic, scanning mirrors are not. So, if the band of wavelengths, which you use during your experiments, is broad, the AOD would be a pretty poor idea, I fear. Best wishes, Johannes On 2018-07-26 03:03, Power, Rory wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your > posting. > ***** > > Hi Confocalists, > > We’re trying to circumvent the annoyance of resonant mirrors screaming > at 8 kHz. I was wondering whether anyone has a recommendation for > constructing soundproof containment? > > Generally were using inexpensive resonant mirrors from EOPC: > http://www.eopc.com/sc30.html > > Also, does anyone know whether the more costly resonant scanners used > for confocal/multi-photon laser scanning microscopies are any better > in this regard? Were not hugely concerned with other attributes > (simply for light sheet microscopy stripe suppression) just the > piercing noise. > > Many thanks in advance! > > Rory > > ………………………………………………………… > > Dr. Rory Power > Morgridge Institute for Research > Medical Engineering | Huisken Lab > 330 N Orchard St, Madison, WI, 53715 > Tel: 608 316 4554 > [hidden email]<mailto:[hidden email]> ... [show rest of quote] -- P. Johannes Helm Voice: (+47) 228 51159 (office) Fax: (+47) 228 51499 (office) |
![]() ![]() |
Martin Wessendorf-2 |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Power, Rory
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear all-- I'm going to guess that the obnoxiousness of the sound may be a function, in part, of the age of the listener. See: https://link.springer.com/article/10.1007/s00420-014-0951-8 --Along those lines, I once had a digital watch with an unusually high-pitched alarm that I (at age 60) could barely hear, but the sound of which drove my sons (in their 20s) crazy. Martin Wessendorf On 7/25/2018 9:33 PM, Power, Rory wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images onhttp://www.imgur.com and include the link in your posting. > ***** > > Hi Mike, > > Thanks for your reply. I was rather referring to an 8 kHz fundamental oscillation. The noise is quite unpleasant - I’d hate to hear the 4 kHz if that’s worse as you say. > > Isolation should be possible. Perhaps combined with some foam we’ll arrive at a workable system. > > Yes, I assumed the same. Thanks again. > > Regards > > Rory > > ………………………………………………………… > > Dr. Rory Power > HFSP Postdoctoral Fellow > Morgridge Institute for Research > Medical Engineering | Huisken Lab > 330 N Orchard St, Madison, WI, 53715 > Tel: 608 316 4554 > [hidden email]<mailto:[hidden email]> > > > > > On Jul 25, 2018, at 21:18, Michael Giacomelli <[hidden email]<mailto:[hidden email]>> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images onhttp://www.imgur.com and include the link in your posting. > ***** > > Hi Rory, > > 8 KHz unidirectional or bidirectional (so 4 KHz fundamental)? The 8 KHz > scanners are usually not that bad even at maximum amplitude because the 8 > kHz tone is a lot harder to hear and essentially all the harmonics are out > of audible range. > > For 4 KHz at larger deflection angles you either need hearing protection or > to really isolate it from the room. A metal box around the scanner and a > window on the input port so that air cannot pass in/out of the scanner > helps a lot. This is how most commerical systems work. > > I don't think the model matters; the noise comes from pushing air, so for a > given mirror size and deflection angle the noise power is going to be > almost constant. > > Mike > > On Wed, Jul 25, 2018 at 9:03 PM, Power, Rory<[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images onhttp://www.imgur.com and include the link in your posting. > ***** > > Hi Confocalists, > > We’re trying to circumvent the annoyance of resonant mirrors screaming at > 8 kHz. I was wondering whether anyone has a recommendation for constructing > soundproof containment? > > Generally were using inexpensive resonant mirrors from EOPC: > http://www.eopc.com/sc30.html > > Also, does anyone know whether the more costly resonant scanners used for > confocal/multi-photon laser scanning microscopies are any better in this > regard? Were not hugely concerned with other attributes (simply for light > sheet microscopy stripe suppression) just the piercing noise. > > Many thanks in advance! > > Rory > > ………………………………………………………… > > Dr. Rory Power > Morgridge Institute for Research > Medical Engineering | Huisken Lab > 330 N Orchard St, Madison, WI, 53715 > Tel: 608 316 4554 > [hidden email]<mailto:[hidden email]> > > > > > > ... [show rest of quote] -- Martin Wessendorf, Ph.D. office: (612) 626-0145 Assoc Prof, Dept Neuroscience lab: (612) 624-2991 University of Minnesota Preferred FAX: (612) 624-8118 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 Minneapolis, MN 55455 e-mail:[hidden email] |
![]() ![]() |
Zdenek Svindrych-2 |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Power, Rory
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Rory, have you considered a regular (non-resonant) galvo for the stripe removal? It should still be fast enough to do more than several round-trips during a 10 ms exposure, so it should be quite effective in stripe removal even without fancy synchronization between the galvo and the camera (depending on the mirror size and the deflection angle you need in your setup...). Best, zdenek -- Zdenek Svindrych, Ph.D. Research Associate - Imaging Specialist Department of Biochemistry and Cell Biology Geisel School of Medicine at Dartmouth email: [hidden email] ---------- Původní e-mail ---------- Od: Power, Rory <[hidden email]> Komu: [hidden email] Datum: 25. 7. 2018 21:21:57 Předmět: Resonant mirror noise "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Confocalists, We’re trying to circumvent the annoyance of resonant mirrors screaming at 8 kHz. I was wondering whether anyone has a recommendation for constructing soundproof containment? Generally were using inexpensive resonant mirrors from EOPC: http://www. eopc.com/sc30.html Also, does anyone know whether the more costly resonant scanners used for confocal/multi-photon laser scanning microscopies are any better in this regard? Were not hugely concerned with other attributes (simply for light sheet microscopy stripe suppression) just the piercing noise. Many thanks in advance! Rory ………………………………………………………… Dr. Rory Power Morgridge Institute for Research Medical Engineering | Huisken Lab 330 N Orchard St, Madison, WI, 53715 Tel: 608 316 4554 [hidden email]<mailto:[hidden email]> " |
![]() ![]() |
Craig Brideau |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Martin Wessendorf-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** To add to the chorus, basically the scanners need to be in a sealed enclosure. It doesn't need to be in vacuum, (although that would really help) but as long as it cannot transfer sound pressure easily via air it will be considerably quieter. Note your scan lens becomes the window for the enclosure; it has quite a bit of spaced glass so it will probably be a good sound absorber. A little sound dampening foam may still be necessary on the exterior, just make sure you can still safely dissipate any heat from the scanners and don't smother the thing completely. I have an early resonant confocal system that shrieks when in use, and a system which is newer by a few years, and the newer one is much quieter. On observation, the new system is simply boxed up better with thicker walls and no gaps from the interior to the exterior. Craig On Thu, Jul 26, 2018 at 8:53 AM Martin Wessendorf <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear all-- > > I'm going to guess that the obnoxiousness of the sound may be a > function, in part, of the age of the listener. > See: https://link.springer.com/article/10.1007/s00420-014-0951-8 > > --Along those lines, I once had a digital watch with an unusually > high-pitched alarm that I (at age 60) could barely hear, but the sound > of which drove my sons (in their 20s) crazy. > > Martin Wessendorf > > > > > On 7/25/2018 9:33 PM, Power, Rory wrote: > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images onhttp://www.imgur.com and include the link in your > posting. > > ***** > > > > Hi Mike, > > > > Thanks for your reply. I was rather referring to an 8 kHz fundamental > oscillation. The noise is quite unpleasant - I’d hate to hear the 4 kHz if > that’s worse as you say. > > > > Isolation should be possible. Perhaps combined with some foam we’ll > arrive at a workable system. > > > > Yes, I assumed the same. Thanks again. > > > > Regards > > > > Rory > > > > ………………………………………………………… > > > > Dr. Rory Power > > HFSP Postdoctoral Fellow > > Morgridge Institute for Research > > Medical Engineering | Huisken Lab > > 330 N Orchard St, Madison, WI, 53715 > > Tel: 608 316 4554 > > [hidden email]<mailto:[hidden email]> > > > > > > > > > > On Jul 25, 2018, at 21:18, Michael Giacomelli <[hidden email]<mailto: > [hidden email]>> wrote: > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images onhttp://www.imgur.com and include the link in your > posting. > > ***** > > > > Hi Rory, > > > > 8 KHz unidirectional or bidirectional (so 4 KHz fundamental)? The 8 KHz > > scanners are usually not that bad even at maximum amplitude because the 8 > > kHz tone is a lot harder to hear and essentially all the harmonics are > out > > of audible range. > > > > For 4 KHz at larger deflection angles you either need hearing protection > or > > to really isolate it from the room. A metal box around the scanner and a > > window on the input port so that air cannot pass in/out of the scanner > > helps a lot. This is how most commerical systems work. > > > > I don't think the model matters; the noise comes from pushing air, so > for a > > given mirror size and deflection angle the noise power is going to be > > almost constant. > > > > Mike > > > > On Wed, Jul 25, 2018 at 9:03 PM, Power, Rory<[hidden email]> > wrote: > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images onhttp://www.imgur.com and include the link in your > posting. > > ***** > > > > Hi Confocalists, > > > > We’re trying to circumvent the annoyance of resonant mirrors screaming at > > 8 kHz. I was wondering whether anyone has a recommendation for > constructing > > soundproof containment? > > > > Generally were using inexpensive resonant mirrors from EOPC: > > http://www.eopc.com/sc30.html > > > > Also, does anyone know whether the more costly resonant scanners used for > > confocal/multi-photon laser scanning microscopies are any better in this > > regard? Were not hugely concerned with other attributes (simply for light > > sheet microscopy stripe suppression) just the piercing noise. > > > > Many thanks in advance! > > > > Rory > > > > ………………………………………………………… > > > > Dr. Rory Power > > Morgridge Institute for Research > > Medical Engineering | Huisken Lab > > 330 N Orchard St, Madison, WI, 53715 > > Tel: 608 316 4554 > > [hidden email]<mailto:[hidden email]> > > > > > > > > > > > > > > -- > Martin Wessendorf, Ph.D. office: (612) 626-0145 > Assoc Prof, Dept Neuroscience lab: (612) 624-2991 > University of Minnesota Preferred FAX: (612) 624-8118 > 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 > Minneapolis, MN 55455 e-mail:[hidden email] > ... [show rest of quote]
|
![]() ![]() |
Power, Rory |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Zdenek Svindrych-2
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, Thanks for the suggestion. We have also used galvos but in this case we are running the camera at significantly higher speed (ca. 400 fps, 2.5 ms exposure) and generally find that the galvos do not like to go above a few 100 Hz (depending on mirror size) for an appreciable scan range. More usually we are using resonant mirrors at ca. 1 kHz, which is sufficient for 100 fps. Regards Rory ………………………………………………………… Dr. Rory Power HFSP Postdoctoral Fellow Morgridge Institute for Research Medical Engineering | Huisken Lab 330 N Orchard St, Madison, WI, 53715 Tel: 608 316 4554 [hidden email]<mailto:[hidden email]> On Jul 26, 2018, at 10:49, [hidden email]<mailto:[hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com<http://www.imgur.com/> and include the link in your posting. ***** Hi Rory, have you considered a regular (non-resonant) galvo for the stripe removal? It should still be fast enough to do more than several round-trips during a 10 ms exposure, so it should be quite effective in stripe removal even without fancy synchronization between the galvo and the camera (depending on the mirror size and the deflection angle you need in your setup...). Best, zdenek -- Zdenek Svindrych, Ph.D. Research Associate - Imaging Specialist Department of Biochemistry and Cell Biology Geisel School of Medicine at Dartmouth email: [hidden email]<mailto:[hidden email]> ---------- Původní e-mail ---------- Od: Power, Rory <[hidden email]<mailto:[hidden email]>> Komu: [hidden email]<mailto:[hidden email]> Datum: 25. 7. 2018 21:21:57 Předmět: Resonant mirror noise "***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Confocalists, We’re trying to circumvent the annoyance of resonant mirrors screaming at 8 kHz. I was wondering whether anyone has a recommendation for constructing soundproof containment? Generally were using inexpensive resonant mirrors from EOPC: http://www. eopc.com/sc30.html<http://eopc.com/sc30.html> Also, does anyone know whether the more costly resonant scanners used for confocal/multi-photon laser scanning microscopies are any better in this regard? Were not hugely concerned with other attributes (simply for light sheet microscopy stripe suppression) just the piercing noise. Many thanks in advance! Rory ………………………………………………………… Dr. Rory Power Morgridge Institute for Research Medical Engineering | Huisken Lab 330 N Orchard St, Madison, WI, 53715 Tel: 608 316 4554 [hidden email]<mailto:[hidden email]><mailto:[hidden email]> " |
![]() ![]() |
Power, Rory |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
In reply to this post by Craig Brideau
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Thanks, yes this seems to be the general consensus. I believe we can build something sealed with the scan lens after the resonant mirror acting as a window. Regards Rory ………………………………………………………… Dr. Rory Power HFSP Postdoctoral Fellow Morgridge Institute for Research Medical Engineering | Huisken Lab 330 N Orchard St, Madison, WI, 53715 Tel: 608 316 4554 [hidden email]<mailto:[hidden email]> On Jul 26, 2018, at 10:55, Craig Brideau <[hidden email]<mailto:[hidden email]>> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** To add to the chorus, basically the scanners need to be in a sealed enclosure. It doesn't need to be in vacuum, (although that would really help) but as long as it cannot transfer sound pressure easily via air it will be considerably quieter. Note your scan lens becomes the window for the enclosure; it has quite a bit of spaced glass so it will probably be a good sound absorber. A little sound dampening foam may still be necessary on the exterior, just make sure you can still safely dissipate any heat from the scanners and don't smother the thing completely. I have an early resonant confocal system that shrieks when in use, and a system which is newer by a few years, and the newer one is much quieter. On observation, the new system is simply boxed up better with thicker walls and no gaps from the interior to the exterior. Craig On Thu, Jul 26, 2018 at 8:53 AM Martin Wessendorf <[hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear all-- I'm going to guess that the obnoxiousness of the sound may be a function, in part, of the age of the listener. See: https://link.springer.com/article/10.1007/s00420-014-0951-8 --Along those lines, I once had a digital watch with an unusually high-pitched alarm that I (at age 60) could barely hear, but the sound of which drove my sons (in their 20s) crazy. Martin Wessendorf On 7/25/2018 9:33 PM, Power, Rory wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images onhttp://www.imgur.com and include the link in your posting. ***** Hi Mike, Thanks for your reply. I was rather referring to an 8 kHz fundamental oscillation. The noise is quite unpleasant - I’d hate to hear the 4 kHz if that’s worse as you say. Isolation should be possible. Perhaps combined with some foam we’ll arrive at a workable system. Yes, I assumed the same. Thanks again. Regards Rory ………………………………………………………… Dr. Rory Power HFSP Postdoctoral Fellow Morgridge Institute for Research Medical Engineering | Huisken Lab 330 N Orchard St, Madison, WI, 53715 Tel: 608 316 4554 [hidden email]<mailto:[hidden email]> On Jul 25, 2018, at 21:18, Michael Giacomelli <[hidden email]<mailto: [hidden email]>> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images onhttp://www.imgur.com and include the link in your posting. ***** Hi Rory, 8 KHz unidirectional or bidirectional (so 4 KHz fundamental)? The 8 KHz scanners are usually not that bad even at maximum amplitude because the 8 kHz tone is a lot harder to hear and essentially all the harmonics are out of audible range. For 4 KHz at larger deflection angles you either need hearing protection or to really isolate it from the room. A metal box around the scanner and a window on the input port so that air cannot pass in/out of the scanner helps a lot. This is how most commerical systems work. I don't think the model matters; the noise comes from pushing air, so for a given mirror size and deflection angle the noise power is going to be almost constant. Mike On Wed, Jul 25, 2018 at 9:03 PM, Power, Rory<[hidden email]> wrote: ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images onhttp://www.imgur.com and include the link in your posting. ***** Hi Confocalists, We’re trying to circumvent the annoyance of resonant mirrors screaming at 8 kHz. I was wondering whether anyone has a recommendation for constructing soundproof containment? Generally were using inexpensive resonant mirrors from EOPC: http://www.eopc.com/sc30.html Also, does anyone know whether the more costly resonant scanners used for confocal/multi-photon laser scanning microscopies are any better in this regard? Were not hugely concerned with other attributes (simply for light sheet microscopy stripe suppression) just the piercing noise. Many thanks in advance! Rory ………………………………………………………… Dr. Rory Power Morgridge Institute for Research Medical Engineering | Huisken Lab 330 N Orchard St, Madison, WI, 53715 Tel: 608 316 4554 [hidden email]<mailto:[hidden email]> -- Martin Wessendorf, Ph.D. office: (612) 626-0145 Assoc Prof, Dept Neuroscience lab: (612) 624-2991 University of Minnesota Preferred FAX: (612) 624-8118 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 Minneapolis, MN 55455 e-mail:[hidden email] |
![]() ![]() |
Craig Brideau |
Loading... |
Reply to author |
Edit post |
Move post |
Delete this post |
Delete this post and replies |
Change post date |
Print post |
Permalink |
Raw mail |
![]() |
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Expanded PVC panel over thin metal plate with a little spacer between them might do it. The Exp PVC is not very sturdy so you will want thin metal plate for structural support, but the foamed nature of the expanded plastic makes it a decent sound absorber without being a linting/dust hazard like acoustic foam. Best of luck! Craig On Thu, Jul 26, 2018 at 11:56 AM Power, Rory <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Thanks, yes this seems to be the general consensus. I believe we can build > something sealed with the scan lens after the resonant mirror acting as a > window. > > Regards > > Rory > > ………………………………………………………… > > Dr. Rory Power > HFSP Postdoctoral Fellow > Morgridge Institute for Research > Medical Engineering | Huisken Lab > 330 N Orchard St, Madison, WI, 53715 > Tel: 608 316 4554 > [hidden email]<mailto:[hidden email]> > > > > > On Jul 26, 2018, at 10:55, Craig Brideau <[hidden email]<mailto: > [hidden email]>> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > To add to the chorus, basically the scanners need to be in a sealed > enclosure. It doesn't need to be in vacuum, (although that would really > help) but as long as it cannot transfer sound pressure easily via air it > will be considerably quieter. Note your scan lens becomes the window for > the enclosure; it has quite a bit of spaced glass so it will probably be a > good sound absorber. A little sound dampening foam may still be necessary > on the exterior, just make sure you can still safely dissipate any heat > from the scanners and don't smother the thing completely. I have an early > resonant confocal system that shrieks when in use, and a system which is > newer by a few years, and the newer one is much quieter. On observation, > the new system is simply boxed up better with thicker walls and no gaps > from the interior to the exterior. > > Craig > > On Thu, Jul 26, 2018 at 8:53 AM Martin Wessendorf <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear all-- > > I'm going to guess that the obnoxiousness of the sound may be a > function, in part, of the age of the listener. > See: https://link.springer.com/article/10.1007/s00420-014-0951-8 > > --Along those lines, I once had a digital watch with an unusually > high-pitched alarm that I (at age 60) could barely hear, but the sound > of which drove my sons (in their 20s) crazy. > > Martin Wessendorf > > > > > On 7/25/2018 9:33 PM, Power, Rory wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images onhttp://www.imgur.com and include the link in your > posting. > ***** > > Hi Mike, > > Thanks for your reply. I was rather referring to an 8 kHz fundamental > oscillation. The noise is quite unpleasant - I’d hate to hear the 4 kHz if > that’s worse as you say. > > Isolation should be possible. Perhaps combined with some foam we’ll > arrive at a workable system. > > Yes, I assumed the same. Thanks again. > > Regards > > Rory > > ………………………………………………………… > > Dr. Rory Power > HFSP Postdoctoral Fellow > Morgridge Institute for Research > Medical Engineering | Huisken Lab > 330 N Orchard St, Madison, WI, 53715 > Tel: 608 316 4554 > [hidden email]<mailto:[hidden email]> > > > > > On Jul 25, 2018, at 21:18, Michael Giacomelli <[hidden email]<mailto: > [hidden email]>> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images onhttp://www.imgur.com and include the link in your > posting. > ***** > > Hi Rory, > > 8 KHz unidirectional or bidirectional (so 4 KHz fundamental)? The 8 KHz > scanners are usually not that bad even at maximum amplitude because the 8 > kHz tone is a lot harder to hear and essentially all the harmonics are > out > of audible range. > > For 4 KHz at larger deflection angles you either need hearing protection > or > to really isolate it from the room. A metal box around the scanner and a > window on the input port so that air cannot pass in/out of the scanner > helps a lot. This is how most commerical systems work. > > I don't think the model matters; the noise comes from pushing air, so > for a > given mirror size and deflection angle the noise power is going to be > almost constant. > > Mike > > On Wed, Jul 25, 2018 at 9:03 PM, Power, Rory<[hidden email]> > wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images onhttp://www.imgur.com and include the link in your > posting. > ***** > > Hi Confocalists, > > We’re trying to circumvent the annoyance of resonant mirrors screaming at > 8 kHz. I was wondering whether anyone has a recommendation for > constructing > soundproof containment? > > Generally were using inexpensive resonant mirrors from EOPC: > http://www.eopc.com/sc30.html > > Also, does anyone know whether the more costly resonant scanners used for > confocal/multi-photon laser scanning microscopies are any better in this > regard? Were not hugely concerned with other attributes (simply for light > sheet microscopy stripe suppression) just the piercing noise. > > Many thanks in advance! > > Rory > > ………………………………………………………… > > Dr. Rory Power > Morgridge Institute for Research > Medical Engineering | Huisken Lab > 330 N Orchard St, Madison, WI, 53715 > Tel: 608 316 4554 > [hidden email]<mailto:[hidden email]> > > > > > > > > -- > Martin Wessendorf, Ph.D. office: (612) 626-0145 > Assoc Prof, Dept Neuroscience lab: (612) 624-2991 > University of Minnesota Preferred FAX: (612) 624-8118 > 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 > Minneapolis, MN 55455 e-mail:[hidden email] > > > ... [show rest of quote]
|