Round coverslips for immunostaining

classic Classic list List threaded Threaded
♦
♦
10 messages Options
Dries Vercauteren Dries Vercauteren
Reply | Threaded
Open this post in threaded view
♦
♦
|

Round coverslips for immunostaining

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Dear all,

I'd need round coverslips for immunostaining cells in wells of 24-well plates. Therefore, I was looking for appropriate coverslips at Menzel. But as you can see
http://www.menzel.de/Cover_Slips.675.0.html?L=1, Menzel has only coverslips of class 1, instead I'd need a 0.17mm thickness (which is best approximated in class 1.5).
Thus, my questions are;
  • what is the ideal diameter of the round coverslips to place them in 24-wells? 12, 13 or 14 mm?
  • Is there another possibility (another company) to buy high quality coverslips of #1.5?
Many thanks in advance for the advice!

Take care,

Dries.
--
Dries Vercauteren, PhD student
Master of Bioscience Engineering: Cell and Gene Biotechnology

Ghent Research Group on Nanomedicines
www.ugent.be/fw/en/research/biofys
Faculty of pharmaceutical sciences, Ghent University
Harelbekestraat 72, 9000 Ghent
Belgium

Phone: +329/264 80 49
Mobile: +32485/30 69 80
E-mail: [hidden email]
[hidden email]
Florian Eich Florian Eich
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Round coverslips for immunostaining

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Dries,
during my PhD I had good experience with the 24-well plates with glass bottom
from http://www.glass-bottom-dishes.com/ (no comercial interest).
If you want to have individual coverslips try http://www.knittel-glaeser.de/ 
(again no com. interest).
They will produce custom diameters if you want.
Regarding the diameter I found it better to use 12mm, giving you a bit of
space to lift the coverslip out of the well.
Best
Florian
Sylvie Le Guyader-2 Sylvie Le Guyader-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Round coverslips for immunostaining

In reply to this post by Dries Vercauteren
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Dries

 

We routinely use 12mm #1.5 coverslips for 24-well plates. You can find them at Warner Instruments.

http://www.warneronline.com/product_info.cfm?id=11&alpha=a&name=Coverslip%2FCoverglass

 

 

Med vänlig hälsning / Best regards

 

Sylvie

 

@@@@@@@@@@@@@@@@@@@@@@@@

Sylvie Le Guyader

Dept of Biosciences and Nutrition

Karolinska Institutet

Novum

14157 Huddinge

Sweden

+46 (0)8 608 9269


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Dries Vercauteren
Sent: 22 August 2008 10:53
To: [hidden email]
Subject: Round coverslips for immunostaining

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear all,

I'd need round coverslips for immunostaining cells in wells of 24-well plates. Therefore, I was looking for appropriate coverslips at Menzel. But as you can see
http://www.menzel.de/Cover_Slips.675.0.html?L=1, Menzel has only coverslips of class 1, instead I'd need a 0.17mm thickness (which is best approximated in class 1.5).
Thus, my questions are;

  • what is the ideal diameter of the round coverslips to place them in 24-wells? 12, 13 or 14 mm?
  • Is there another possibility (another company) to buy high quality coverslips of #1.5?

Many thanks in advance for the advice!

Take care,

Dries.
--
Dries Vercauteren, PhD student
Master of Bioscience Engineering: Cell and Gene Biotechnology

Ghent Research Group on Nanomedicines
www.ugent.be/fw/en/research/biofys
Faculty of pharmaceutical sciences, Ghent University
Harelbekestraat 72, 9000 Ghent
Belgium

Phone: +329/264 80 49
Mobile: +32485/30 69 80
E-mail: [hidden email]
[hidden email]

Beat Ludin Beat Ludin
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Round coverslips for immunostaining

In reply to this post by Dries Vercauteren
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Dries

I would recommend the thickness-selected 0.17mm (=type 1014)
coverslips from "assistent" (http://www.hecht-assistent.de). Use 12mm
coverslips for 24-well plates. Larger coverslips are difficult to get
out of the wells.
No commercial interest.

Beat


At 10:53 22-08-2008, you wrote:

>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>Dear all,
>
>I'd need round coverslips for immunostaining cells in wells of
>24-well plates. Therefore, I was looking for appropriate coverslips
>at Menzel. But as you can see
><http://www.menzel.de/Cover_Slips.675.0.html?L=1>http://www.menzel.de/Cover_Slips.675.0.html?L=1,
>Menzel has only coverslips of class 1, instead I'd need a 0.17mm
>thickness (which is best approximated in class 1.5).
>Thus, my questions are;
>    * what is the ideal diameter of the round coverslips to place
> them in 24-wells? 12, 13 or 14 mm?
>    * Is there another possibility (another company) to buy high
> quality coverslips of #1.5?
>Many thanks in advance for the advice!
>
>Take care,
>
>Dries.
>--
>Dries Vercauteren, PhD student
>Master of Bioscience Engineering: Cell and Gene Biotechnology
>
>Ghent Research Group on Nanomedicines
><http://www.ugent.be/fw/en/research/biofys>www.ugent.be/fw/en/research/biofys
>Faculty of pharmaceutical sciences, Ghent University
>Harelbekestraat 72, 9000 Ghent
>Belgium
>
>Phone: +329/264 80 49
>Mobile: +32485/30 69 80
>E-mail: [hidden email]
><mailto:[hidden email]>[hidden email]
RICHARD BURRY RICHARD BURRY
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Round coverslips for immunostaining

In reply to this post by Dries Vercauteren
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal  
Dries
 
Yes, 12 mm is the best, because anything bigger is impossible to get out of the well.  12 mm gives a good amount of area for growing cells, anything small well limit the number of cells.  Please experiment with ways to remover the cover slip and still know which side has the cells!  Until mastered this was our only problem.
 
Dick
 
R. W. Burry, Campus Microscopy and Imaging Facility, Director
The Ohio State University
277 Biomedical Research Tower
460 West Twelfth Avenue
Columbus, Ohio 43210

----- Original Message -----
From: Dries Vercauteren <[hidden email]>
Date: Friday, August 22, 2008 5:03 am
Subject: Round coverslips for immunostaining
To: [hidden email]

> Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Dear all,

> I'd need round coverslips for immunostaining cells in wells of 24-well plates. Therefore, I was looking for appropriate coverslips at Menzel. But as you can see
> http://www.menzel.de/Cover_Slips.675.0.html?L=1, Menzel has only coverslips of class 1, instead I'd need a 0.17mm thickness (which is best approximated in class 1.5).
> Thus, my questions are;
  • > what is the ideal diameter of the round coverslips to place them in 24-wells? 12, 13 or 14 mm?
  • > Is there another possibility (another company) to buy high quality coverslips of #1.5?
Many thanks in advance for the advice!

> Take care,

> Dries.
> --
> Dries Vercauteren, PhD student
> Master of Bioscience Engineering: Cell and Gene Biotechnology

> Ghent Research Group on Nanomedicines
> www.ugent.be/fw/en/research/biofys
> Faculty of pharmaceutical sciences, Ghent University
> Harelbekestraat 72, 9000 Ghent
> Belgium

> Phone: +329/264 80 49
> Mobile: +32485/30 69 80
> E-mail: [hidden email]
<A href="javascript:main.compose('new','t=driesvercauteren@gmail.com')">> driesvercauteren@...



> Spam
> Not spam
> Forget previous vote





 
Sam's Mail Sam's Mail
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Round coverslips for immunostaining. .

In reply to this post by Beat Ludin
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Does anyone know who distributes this Hecht-assistant 0.17mm thick glass here in the US?

--
Samuel A. Connell
Director of Light Microscopy
Cell & Tissue Imaging Center
St. Jude Children's Research Hospital
262 Danny Thomas Place
Memphis, TN 38105-3678
Office (901) 495-2536
Cell (901) 603-3162
[hidden email]



-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Beat Ludin
Sent: Friday, August 22, 2008 6:04 AM
To: [hidden email]
Subject: Re: Round coverslips for immunostaining. .

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Dries

I would recommend the thickness-selected 0.17mm (=type 1014)
coverslips from "assistent" (http://www.hecht-assistent.de). Use 12mm
coverslips for 24-well plates. Larger coverslips are difficult to get
out of the wells.
No commercial interest.

Beat


At 10:53 22-08-2008, you wrote:

>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>Dear all,
>
>I'd need round coverslips for immunostaining cells in wells of
>24-well plates. Therefore, I was looking for appropriate coverslips
>at Menzel. But as you can see
><http://www.menzel.de/Cover_Slips.675.0.html?L=1>http://www.menzel.de/Cover_Slips.675.0.html?L=1,
>Menzel has only coverslips of class 1, instead I'd need a 0.17mm
>thickness (which is best approximated in class 1.5).
>Thus, my questions are;
>    * what is the ideal diameter of the round coverslips to place
> them in 24-wells? 12, 13 or 14 mm?
>    * Is there another possibility (another company) to buy high
> quality coverslips of #1.5?
>Many thanks in advance for the advice!
>
>Take care,
>
>Dries.
>--
>Dries Vercauteren, PhD student
>Master of Bioscience Engineering: Cell and Gene Biotechnology
>
>Ghent Research Group on Nanomedicines
><http://www.ugent.be/fw/en/research/biofys>www.ugent.be/fw/en/research/biofys
>Faculty of pharmaceutical sciences, Ghent University
>Harelbekestraat 72, 9000 Ghent
>Belgium
>
>Phone: +329/264 80 49
>Mobile: +32485/30 69 80
>E-mail: [hidden email]
><mailto:[hidden email]>[hidden email]
Stephen Cody Stephen Cody
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Round coverslips for immunostaining

In reply to this post by Beat Ludin
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

G'day dries,
 
As an alternative to the coverslips that Beat recommends, you should also be able to order special confocal thickness coverslips from Menzel, by special order in any size you wish. The thicness recommended for confocal, especially if you are going to use a water immersion lens is "Grade: 0.170 +/- 0.01mm". We order larger round coverslips of this thicness through an Australian agency (Lomb), but they custom order them from Menzel.
 
This has been discussed previously on the list, se subject:
"Coverslip - Good News"
 
Cheers
Stephen H. Cody
Microscopy Manager
Central Resource for Advanced Microscopy
Ludwig Institute for Cancer Research
PO Box 2008 Royal Melbourne Hospital
Parkville, Victoria,      3050
Australia
Tel: 61 3 9341 3155    Fax: 61 3 9341 3104
email: [hidden email]
www.ludwig.edu.au/labs/confocal.html
www.ludwig.edu.au/confocal

Tip: Learn how to receive reminders about you microscope booking:
http://www.ludwig.edu.au/confocal/Local/Booking_Hint.htm Type your signature here

________________________________

From: Confocal Microscopy List on behalf of Beat Ludin
Sent: Fri 22/08/2008 9:04 PM
To: [hidden email]
Subject: Re: Round coverslips for immunostaining



Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Dries

I would recommend the thickness-selected 0.17mm (=type 1014)
coverslips from "assistent" (http://www.hecht-assistent.de <http://www.hecht-assistent.de/> ). Use 12mm
coverslips for 24-well plates. Larger coverslips are difficult to get
out of the wells.
No commercial interest.

Beat


At 10:53 22-08-2008, you wrote:

>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>Dear all,
>
>I'd need round coverslips for immunostaining cells in wells of
>24-well plates. Therefore, I was looking for appropriate coverslips
>at Menzel. But as you can see
><http://www.menzel.de/Cover_Slips.675.0.html?L=1>http://www.menzel.de/Cover_Slips.675.0.html?L=1,
>Menzel has only coverslips of class 1, instead I'd need a 0.17mm
>thickness (which is best approximated in class 1.5).
>Thus, my questions are;
>    * what is the ideal diameter of the round coverslips to place
> them in 24-wells? 12, 13 or 14 mm?
>    * Is there another possibility (another company) to buy high
> quality coverslips of #1.5?
>Many thanks in advance for the advice!
>
>Take care,
>
>Dries.
>--
>Dries Vercauteren, PhD student
>Master of Bioscience Engineering: Cell and Gene Biotechnology
>
>Ghent Research Group on Nanomedicines
><http://www.ugent.be/fw/en/research/biofys>www.ugent.be/fw/en/research/biofys
>Faculty of pharmaceutical sciences, Ghent University
>Harelbekestraat 72, 9000 Ghent
>Belgium
>
>Phone: +329/264 80 49
>Mobile: +32485/30 69 80
>E-mail: [hidden email]
><mailto:[hidden email]>[hidden email]
Sylvie Le Guyader-2 Sylvie Le Guyader-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Round coverslips for immunostaining

In reply to this post by RICHARD BURRY
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi again Dries

 

To remove coverslips from wells I use a long needle (of the type to do injections) bent to get just the tip at a 90º angle. I dip the needle tip and a pair of flat tweezers into Ethanol to sterilize them then I scratch the bottom of the well with the needle tip until it catches and lifts the coverslip a bit. I then carefully grab the coverslip with the flat tweezers, making sure I keep the side that has the cells towards the top. You will break a couple of coverslips first then you’ll become an expert and will be able to do it very quickly and reliably.

 

Med vänlig hälsning / Best regards

 

Sylvie

 

@@@@@@@@@@@@@@@@@@@@@@@@

Sylvie Le Guyader

Dept of Biosciences and Nutrition

Karolinska Institutet

Novum

14157 Huddinge

Sweden

+46 (0)8 608 9269


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of RICHARD BURRY
Sent: 22 August 2008 15:47
To: [hidden email]
Subject: Re: Round coverslips for immunostaining

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal  
Dries
 
Yes, 12 mm is the best, because anything bigger is impossible to get out of the well.  12 mm gives a good amount of area for growing cells, anything small well limit the number of cells.  Please experiment with ways to remover the cover slip and still know which side has the cells!  Until mastered this was our only problem.
 
Dick
 
R. W. Burry, Campus Microscopy and Imaging Facility, Director
The Ohio State University
277 Biomedical Research Tower
460 West Twelfth Avenue
Columbus, Ohio 43210

----- Original Message -----
From: Dries Vercauteren <[hidden email]>
Date: Friday, August 22, 2008 5:03 am
Subject: Round coverslips for immunostaining
To: [hidden email]

> Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear all,

> I'd need round coverslips for immunostaining cells in wells of 24-well plates. Therefore, I was looking for appropriate coverslips at Menzel. But as you can see
> http://www.menzel.de/Cover_Slips.675.0.html?L=1, Menzel has only coverslips of class 1, instead I'd need a 0.17mm thickness (which is best approximated in class 1.5).
> Thus, my questions are;

  • > what is the ideal diameter of the round coverslips to place them in 24-wells? 12, 13 or 14 mm?
  • > Is there another possibility (another company) to buy high quality coverslips of #1.5?

Many thanks in advance for the advice!

> Take care,

> Dries.
> --
> Dries Vercauteren, PhD student
> Master of Bioscience Engineering: Cell and Gene Biotechnology

> Ghent Research Group on Nanomedicines
> www.ugent.be/fw/en/research/biofys
> Faculty of pharmaceutical sciences, Ghent University
> Harelbekestraat 72, 9000 Ghent
> Belgium

> Phone: +329/264 80 49
> Mobile: +32485/30 69 80
> E-mail: [hidden email]
<a href="javascript:main.compose('new','t=driesvercauteren@gmail.com')">> driesvercauteren@...



> Spam
> Not spam
> Forget previous vote





 

Dries Vercauteren Dries Vercauteren
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Round coverslips for immunostaining

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Dear all,

many thanks to anyone of you for all these helpful comments!

Take care!

Dries.


2008/8/25 Sylvie Le Guyader <[hidden email]>

Hi again Dries

 

To remove coverslips from wells I use a long needle (of the type to do injections) bent to get just the tip at a 90º angle. I dip the needle tip and a pair of flat tweezers into Ethanol to sterilize them then I scratch the bottom of the well with the needle tip until it catches and lifts the coverslip a bit. I then carefully grab the coverslip with the flat tweezers, making sure I keep the side that has the cells towards the top. You will break a couple of coverslips first then you'll become an expert and will be able to do it very quickly and reliably.

 

Med vänlig hälsning / Best regards

 

Sylvie

 

@@@@@@@@@@@@@@@@@@@@@@@@

Sylvie Le Guyader

Dept of Biosciences and Nutrition

Karolinska Institutet

Novum

14157 Huddinge

Sweden

+46 (0)8 608 9269


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of RICHARD BURRY
Sent: 22 August 2008 15:47

Subject: Re: Round coverslips for immunostaining

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal  


Dries
 
Yes, 12 mm is the best, because anything bigger is impossible to get out of the well.  12 mm gives a good amount of area for growing cells, anything small well limit the number of cells.  Please experiment with ways to remover the cover slip and still know which side has the cells!  Until mastered this was our only problem.
 
Dick
 
R. W. Burry, Campus Microscopy and Imaging Facility, Director
The Ohio State University
277 Biomedical Research Tower
460 West Twelfth Avenue
Columbus, Ohio 43210

----- Original Message -----
From: Dries Vercauteren <[hidden email]>
Date: Friday, August 22, 2008 5:03 am
Subject: Round coverslips for immunostaining
To: [hidden email]

Dear all,

> I'd need round coverslips for immunostaining cells in wells of 24-well plates. Therefore, I was looking for appropriate coverslips at Menzel. But as you can see
> http://www.menzel.de/Cover_Slips.675.0.html?L=1, Menzel has only coverslips of class 1, instead I'd need a 0.17mm thickness (which is best approximated in class 1.5).
> Thus, my questions are;

  • > what is the ideal diameter of the round coverslips to place them in 24-wells? 12, 13 or 14 mm?
  • > Is there another possibility (another company) to buy high quality coverslips of #1.5?

Many thanks in advance for the advice!

> Take care,

> Dries.
> --
> Dries Vercauteren, PhD student
> Master of Bioscience Engineering: Cell and Gene Biotechnology

> Ghent Research Group on Nanomedicines
> www.ugent.be/fw/en/research/biofys
> Faculty of pharmaceutical sciences, Ghent University
> Harelbekestraat 72, 9000 Ghent
> Belgium

> Phone: +329/264 80 49
> Mobile: +32485/30 69 80
> E-mail: [hidden email]
> [hidden email]



> Spam
> Not spam
> Forget previous vote





 




Jose Paulo Sequeira Farinha Jose Paulo Sequeira Farinha
Reply | Threaded
Open this post in threaded view
♦
♦
|

Open Confocal/Multiphoton/FCS research position in Lisbon, Portugal

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Main research field: Chemistry

Sub research field: Fluorescence Spectroscopy and Imaging

Job description: The researcher will develop an independent project in  
Fluorescence
Spectroscopy and Imaging, specifically in Confocal/Biphotonic  
Fluorescence Microscopy and
Fluorescence Correlation Spectroscopy. Centro de Química-Física  
Molecular (CQFM,
http://dequim.ist.utl.pt/CQFM) is part of IN - Institute of Nanoscience and
Nanotechnology, and a research center of Instituto Superior Técnico,  
one of the oldest
and largest Schools of Science and Engineering in Portugal
Duties of successful candidate include responsibility for a research  
project to be
developed within the unit. There will be no teaching duties. The  
working language is
English.

The successful candidates will receive a salary in accordance with the  
university
regulations for a senior researcher. The contract offered will have  
duration of up to 5
years, renewed yearly based on mutual agreement. The annual gross  
income, before taxes,
will be approximately 43,000 Euros.

Applicants
Applicants should have obtained a Ph. D. in one of the following  
areas: Chemistry,
Physics, Biochemistry, Chemical Engineering, Biological Engineering  
and Materials.
Candidates should have a high quality research record and research  
experience. The
successful candidates are expected to do research in the areas listed  
above. For more
information, consult http://www.eracareers.pt

How to Apply
By email to: [hidden email], with the following information:
- Identification of the candidate
- Curriculum Vitae/resume
- Letters of Reference
- Statement of purpose for the period of the contract (up to 3 pages).


Good weather is bonus...