Hello,
 
I'm looking for information about practical experiences with the SP5 resonant scanner.
 
Leica is claiming: resonant scanners have a much better signal/noise when compared with normal scanning devices, but still offer true optical sectioning at the diffraction limit.
 
Do you have any practical evidence on this?
 
Thanks for your help.
 
Dimitri
 


Envoyé avec Yahoo! Mail.
Une boite mail plus intelligente.
Craig Brideau Craig Brideau
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: SP5 resonant scanner

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal All resonant scanners really do is go faster.  They do have smoother motion because they are operating at a mechanical resonance point, but they are not some sort of magic bullet.  They just sweep the laser faster, which if you have a weak sample will require you to turn up your laser as the sample will have a shorter effective exposure time.

Craig


On Mon, Apr 7, 2008 at 7:03 AM, Gurrie Dimitri <[hidden email]> wrote:
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Hello,
 
I'm looking for information about practical experiences with the SP5 resonant scanner.
 
Leica is claiming: resonant scanners have a much better signal/noise when compared with normal scanning devices, but still offer true optical sectioning at the diffraction limit.
 
Do you have any practical evidence on this?
 
Thanks for your help.
 
Dimitri
 


Envoyé avec Yahoo! Mail.
Une boite mail plus intelligente.

Nuno Moreno Nuno Moreno
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: SP5 resonant scanner

In reply to this post by Gurrie Dimitri
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

forget the better do signal/noise ratio. Actually because pixel dwell
time is short, the pmts have less time to make a more statistically
accurate measurement. What leica claims is that due to faster scanning,
the number of molecules in the exited state and triples which get one
more photon is decreased, consequently decreasing bleaching. As far as I
remember, the resonant scanner operates at 8000 lps which makes a pixel
occurring in the range of tens of ns. If you remember a typical
relaxation half life time is <10ns, meaning that at some degree this can
be true.

Apart from the numbers, we do realize that bleaching is in fact less but
it is not linear (some time because you have to use more laser power its
actually higher). This can be due to the lifetime of the fluophore, due
to environment conditions, etc...

About speed I remember taking pictures at 3ms frame rate with a 16 pixel
height, having the complete confocality.

I had never try it but I would say that the sp5x (fiber white laser) can
have a laser power limitation when using the resonant head.

No commercial interest,
NM



Gurrie Dimitri wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
> Hello,
>  
> I'm looking for information about practical experiences with the SP5
> resonant scanner.
>  
> Leica is claiming: resonant scanners have a much better signal/noise
> when compared with normal scanning devices, but still offer true optical
> sectioning at the diffraction limit.
>  
> Do you have any practical evidence on this?
>  
> Thanks for your help.
>  
> Dimitri
>  
>
> ------------------------------------------------------------------------
> Envoyé avec Yahoo! Mail
> <http://us.rd.yahoo.com/mailuk/taglines/isp/control/*http://us.rd.yahoo.com/evt=52423/*http://fr.docs.yahoo.com/mail/overview/index.html>.
> Une boite mail plus intelligente.

--
Nuno Moreno
Cell Imaging Unit
Instituto Gulbenkian de Ciência
http://uic.igc.gulbekian.pt
http://www.igc.gulbekian.pt
phone +351 214464606
fax   +351 214407970
Guy Cox Guy Cox
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: SP5 resonant scanner

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Re: SP5 resonant scanner
Nuno Moreno wrote:
 "What leica claims is that due to faster scanning,
the number of molecules in the exited state and triples which get one
more photon is decreased, consequently decreasing bleaching. As far as I
remember, the resonant scanner operates at 8000 lps which makes a pixel
occurring in the range of tens of ns. If you remember a typical
relaxation half life time is <10ns, meaning that at some degree this can
be true."

I'm not sure I buy this.  If we got down to 1ns dwell time than maybe - but
then we wouldn't get an image because the dwell is shorter than the lifetime. 
In a conventional 4ms dwell time the likelihood of two photons arriving within
one 10ns period is probably lower than with the fast scanner because we
can use a lower laser intensity.  Once the dwell time exceeds the relaxation
time I can't see that this effect could apply.
                                                                                                  Guy

 
Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
     http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
http://www.guycox.net

Nuno Moreno Nuno Moreno
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: SP5 resonant scanner

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Do you think it can have some effect for triplet states?




Guy Cox wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
> Nuno Moreno wrote:
>  "What leica claims is that due to faster scanning,
> the number of molecules in the exited state and triples which get one
> more photon is decreased, consequently decreasing bleaching. As far as I
> remember, the resonant scanner operates at 8000 lps which makes a pixel
> occurring in the range of tens of ns. If you remember a typical
> relaxation half life time is <10ns, meaning that at some degree this can
> be true."
>
> I'm not sure I buy this.  If we got down to 1ns dwell time than maybe - but
> then we wouldn't get an image because the dwell is shorter than the
> lifetime.
> In a conventional 4ms dwell time the likelihood of two photons arriving
> within
> one 10ns period is probably lower than with the fast scanner because we
> can use a lower laser intensity.  Once the dwell time exceeds the
> relaxation
> time I can't see that this effect could apply.
>                                                                                                  
> Guy
>
>  
> Optical Imaging Techniques in Cell Biology
> by Guy Cox    CRC Press / Taylor & Francis
>      http://www.guycox.com/optical.htm
> ______________________________________________
> Associate Professor Guy Cox, MA, DPhil(Oxon)
> Electron Microscope Unit, Madsen Building F09,
> University of Sydney, NSW 2006
> ______________________________________________
> Phone +61 2 9351 3176     Fax +61 2 9351 7682
> Mobile 0413 281 861
> ______________________________________________
> http://www.guycox.net 
> <https://www.mcws.usyd.edu.au/exchweb/bin/redir.asp?URL=https://www.mcws.usyd.edu.au/exchweb/bin/redir.asp?URL=http://www.guycox.net>
>

--
Nuno Moreno
Cell Imaging Unit
Instituto Gulbenkian de Ciência
http://uic.igc.gulbekian.pt
http://www.igc.gulbekian.pt
phone +351 214464606
fax   +351 214407970
James Pawley James Pawley
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: SP5 resonant scanner

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Re: SP5 resonant scanner
Hi all,

Just to muddy the waters a little more...

The extra laser power is needed not only because the frame time is so fast (i,e., because you still need the same number of detected photons to make an image of a given quality) but also because only a small fraction of the a sinusoidal horizontal scan is linear enough to use effectively (You can try to linearize it by skewing the pixel clock, but only so much). This fraction is about 30%, even if you scan in both directions, 15% if you scan in one direction, versus 60-90% for non-resonant scanners. Assuming that the beam is blanked or masked when not scanning the area from which the signal is collected, this still means that the beam power must be 3-6x higher with resonant scanning even if you collect each image for the same time (i.e., if you compare a Kalman average of 25-30 video scans or collect a single, 1s scan).

As far as triplet-state effects however, there may be something to the argument. Triplet states last a lot longer than singlet states: ms to microseconds vs ns.  So the fact that, after the ~30ns pixel dwell time, any triplet excitation will have a chance to decay during the frame time before again being assaulted by photons that might excite it from the triplet state to a state where permanent damage would occur, might be an advantage.

Of course, we need to remember that it isn't quite this simple. Assuming Nyquist sampling, each molecule is actually illuminated for a time equal to 4 pixel-dwell times, and this happens on 4 successive scan lines. Even so, any molecule excited into the triplet state in this way would spend relatively less time under illumination with a resonant, video-rate raster than with a 1 second linear scan (assuming 512x512 raster, 3 microseconds/pixel).

Cheers,

Jim P.
              **********************************************
Prof. James B. Pawley,                                          Ph.  608-263-3147 
Room 223, Zoology Research Building,                                  FAX  608-265-5315
1117 Johnson Ave., Madison, WI, 53706                                [hidden email]
3D Microscopy of Living Cells Course, June 14-26, 2008, UBC, Vancouver Canada
Info: http://www.3dcourse.ubc.ca/            Applications still being accepted
               "If it ain't diffraction, it must be statistics." Anon.
Do you think it can have some effect for triplet states?

Guy Cox wrote:
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Nuno Moreno wrote:
 "What leica claims is that due to faster scanning,
the number of molecules in the exited state and triples which get one
more photon is decreased, consequently decreasing bleaching. As far as I
remember, the resonant scanner operates at 8000 lps which makes a pixel
occurring in the range of tens of ns. If you remember a typical
relaxation half life time is <10ns, meaning that at some degree this can
be true."

I'm not sure I buy this.  If we got down to 1ns dwell time than maybe - but
then we wouldn't get an image because the dwell is shorter than the lifetime. In a conventional 4ms dwell time the likelihood of two photons arriving within
one 10ns period is probably lower than with the fast scanner because we
can use a lower laser intensity.  Once the dwell time exceeds the relaxation
time I can't see that this effect could apply.
                                                                                                 
Guy

 Optical Imaging Techniques in Cell Biology
by Guy Cox    CRC Press / Taylor & Francis
     http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Electron Microscope Unit, Madsen Building F09,
University of Sydney, NSW 2006
______________________________________________
Phone +61 2 9351 3176     Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
http://www.guycox.net <https://www.mcws.usyd.edu.au/exchweb/bin/redir.asp?URL=https://www.mcws.usyd.edu.au/exchweb/bin/redir.asp?URL=http://www.guycox.net>


--
Nuno Moreno
Cell Imaging Unit
Instituto Gulbenkian de Ciência
http://uic.igc.gulbekian.pt
http://www.igc.gulbekian.pt
phone +351 214464606
fax   +351 214407970


--
Martin Hoppe-2 Martin Hoppe-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: SP5 resonant scanner

In reply to this post by Nuno Moreno
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
All,
a summary of the effects is given in the publication below.
Martin
 
MRT Letter: High Speed Scanning Has the Potential to
Increase Fluorescence Yield and to Reduce Photobleaching
Rolf T. Borlinghaus
Microsc. Res. Tech. 69: 2006. Wiley-Liss, Inc.

------------------------------------------------------------------------------
Martin Hoppe, Ph.D.
Head of Market Management & Confocal Applications
Life Science Division
Leica Microsystems CMS GmbH
Am Friedensplatz 3 | 68165 Mannheim (Germany)
Phone : +49 621 7028 1100 | Fax : +49 621 7028 1180
Cell: +49 172 623 0409




-----Ursprüngliche Mitteilung-----
Von: Nuno Moreno <[hidden email]>
An: [hidden email]
Verschickt: Di., 8. Apr. 2008, 16:10
Thema: Re: SP5 resonant scanner

Search the CONFOCAL archive at 
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 
 
Do you think it can have some effect for triplet states? 
 
 
Guy Cox wrote: 
> Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 
> Nuno Moreno wrote: 
> "What leica claims is that due to faster scanning, 
> the number of molecules in the exited state and triples which get one 
> more photon is decreased, consequently decreasing bleaching. As far as I 
> remember, the resonant scanner operates at 8000 lps which makes a pixel 
> occurring in the range of tens of ns. If you remember a typical 
> relaxation half life time is <10ns, meaning that at some degree this can 
> be true." 
> > I'm not sure I buy this. If we got down to 1ns dwell time than maybe - but 
> then we wouldn't get an image because the dwell is shorter than the > lifetime. > In a conventional 4ms dwell time the likelihood of two photons arriving > within 
> one 10ns period is probably lower than with the fast scanner because we 
> can use a lower laser intensity. Once the dwell time exceeds the > relaxation 
> time I can't see that this effect could apply. 
> > Guy 
> > > Optical Imaging Techniques in Cell Biology 
> by Guy Cox CRC Press / Taylor & Francis 
> http://www.guycox.com/optical.htm 
> ______________________________________________ 
> Associate Professor Guy Cox, MA, DPhil(Oxon) 
> Electron Microscope Unit, Madsen Building F09, 
> University of Sydney, NSW 2006 
> ______________________________________________ 
> Phone +61 2 9351 3176 Fax +61 2 9351 7682 
> Mobile 0413 281 861 
> ______________________________________________ 
> http://www.guycox.net > <https://www.mcws.usyd.edu.au/exchweb/bin/redir.asp?URL=https://www.mcws.usyd.edu.au/exchweb/bin/redir.asp?URL=http://www.guycox.net> 
>  
-- Nuno Moreno 
Cell Imaging Unit 
Instituto Gulbenkian de Ciência 
http://uic.igc.gulbekian.pt 
http://www.igc.gulbekian.pt 
phone +351 214464606 
fax +351 214407970 

Bei AOL gibt's jetzt kostenlos eMail für alle! Was es sonst noch umsonst bei AOL gibt, finden Sie hier heraus AOL.de.
Martin Hoppe-2 Martin Hoppe-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: SP5 resonant scanner - complete reference

In reply to this post by Nuno Moreno
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Sorry,

here is the complete reference:

Borlinghaus, R.T.: "High speed scanning has the potential to increase fluorescence yield and to reduce photobleaching"
Microsc. Res. Tech. 2006 Sep; 69(9): pp 689-692.

------------------------------------------------------------------------------
Martin Hoppe, Ph.D.
Head of Market Management & Confocal Applications
Life Science Division
Leica Microsystems CMS GmbH
Am Friedensplatz 3 | 68165 Mannheim (Germany)
Phone : +49 621 7028 1100 | Fax : +49 621 7028 1180
Cell: +49 172 623 0409




-----Ursprüngliche Mitteilung-----
Von: Nuno Moreno <[hidden email]>
An: [hidden email]
Verschickt: Di., 8. Apr. 2008, 16:10
Thema: Re: SP5 resonant scanner

Search the CONFOCAL archive at 
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 
 
Do you think it can have some effect for triplet states? 
 
 
Guy Cox wrote: 
> Search the CONFOCAL archive at > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 
> Nuno Moreno wrote: 
> "What leica claims is that due to faster scanning, 
> the number of molecules in the exited state and triples which get one 
> more photon is decreased, consequently decreasing bleaching. As far as I 
> remember, the resonant scanner operates at 8000 lps which makes a pixel 
> occurring in the range of tens of ns. If you remember a typical 
> relaxation half life time is <10ns, meaning that at some degree this can 
> be true." 
> > I'm not sure I buy this. If we got down to 1ns dwell time than maybe - but 
> then we wouldn't get an image because the dwell is shorter than the > lifetime. > In a conventional 4ms dwell time the likelihood of two photons arriving > within 
> one 10ns period is probably lower than with the fast scanner because we 
> can use a lower laser intensity. Once the dwell time exceeds the > relaxation 
> time I can't see that this effect could apply. 
> > Guy 
> > > Optical Imaging Techniques in Cell Biology 
> by Guy Cox CRC Press / Taylor & Francis 
> http://www.guycox.com/optical.htm 
> ______________________________________________ 
> Associate Professor Guy Cox, MA, DPhil(Oxon) 
> Electron Microscope Unit, Madsen Building F09, 
> University of Sydney, NSW 2006 
> ______________________________________________ 
> Phone +61 2 9351 3176 Fax +61 2 9351 7682 
> Mobile 0413 281 861 
> ______________________________________________ 
> http://www.guycox.net > <https://www.mcws.usyd.edu.au/exchweb/bin/redir.asp?URL=https://www.mcws.usyd.edu.au/exchweb/bin/redir.asp?URL=http://www.guycox.net> 
>  
-- Nuno Moreno 
Cell Imaging Unit 
Instituto Gulbenkian de Ciência 
http://uic.igc.gulbekian.pt 
http://www.igc.gulbekian.pt 
phone +351 214464606 
fax +351 214407970 

Bei AOL gibt's jetzt kostenlos eMail für alle! Was es sonst noch umsonst bei AOL gibt, finden Sie hier heraus AOL.de.
Alberto Diaspro Alberto Diaspro
Reply | Threaded
Open this post in threaded view
♦
♦
|

SOS alby vs. mats gustafsson

In reply to this post by Craig Brideau
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Please, Mats contact me...
Dear folks sorry for this Alby-Mats message ... it seems my e-mails do  
not reach Mats...thank you for comphrension.
All the best from Erice
Alby

----------------------------------------------------
"Water slowly flowed under the sky" (Cesare Pavese)
-----------------------------------------------------
  Alberto Diaspro, EBSA President-Elect, MicroScoBIO LAMBS-IFOM,  
Department of Physics, University of Genoa, Via Dodecaneso 33, 16146  
Genoa, Italy - fax +39-010314218 - tel +39 0103536426/309; URLs: www.lambs.it
;
EBSA is Biophysics in Europe - European Biophysical Societies'  
Association www.ebsa.org
  ----------------------------------------------