Secondary antibody for 405nm laser line ?

classic Classic list List threaded Threaded
♦
♦
22 messages Options
12
lechristophe lechristophe
Reply | Threaded
Open this post in threaded view
♦
♦
|

Secondary antibody for 405nm laser line ?

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear list,

I would like to use all four lines of our new confocal for immunolabeling (405,488,532,633). For the moment I use Hoechst/Alexa488/Alexa555/Cy5, but I would like to have a fourth antibody labeling using the 405 channel instead of Hoechst. What secondary fluorescent antibody would you recommend ? Is there an Alexa405 or something similar that would be bright and stable ?

Thanks a lot,

Christophe Leterrier

Postdoc
INSERM UMR641 Neurobiology of ionic channels
IFR Jean Roche - Mediterranee University
Marseille, France
Claudia Campos Claudia Campos
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Secondary antibody for 405nm laser line ?

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Christophe Leterrier escreveu:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear list,
>
> I would like to use all four lines of our new confocal for
> immunolabeling (405,488,532,633). For the moment I use
> Hoechst/Alexa488/Alexa555/Cy5, but I would like to have a fourth
> antibody labeling using the 405 channel instead of Hoechst. What
> secondary fluorescent antibody would you recommend ? Is there an
> Alexa405 or something similar that would be bright and stable ?
>
> Thanks a lot,
>
> Christophe Leterrier
>
> Postdoc
> INSERM UMR641 Neurobiology of ionic channels
> IFR Jean Roche - Mediterranee University
> Marseille, France

Molecular probes has Alexa 405, but I have never used it.

Cheers

Claudia Gonzalez lab
University of Geneva
Switzerland
Alison J. North Alison J. North
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Secondary antibody for 405nm laser line ?

In reply to this post by lechristophe
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

One of our users tried AlexaFluor 405 with the 405 line of our confocal
and said it was absolutely gorgeous (I quote!).

Best,
Alison

Christophe Leterrier wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear list,
>
> I would like to use all four lines of our new confocal for
> immunolabeling (405,488,532,633). For the moment I use
> Hoechst/Alexa488/Alexa555/Cy5, but I would like to have a fourth
> antibody labeling using the 405 channel instead of Hoechst. What
> secondary fluorescent antibody would you recommend ? Is there an
> Alexa405 or something similar that would be bright and stable ?
>
> Thanks a lot,
>
> Christophe Leterrier
>
> Postdoc
> INSERM UMR641 Neurobiology of ionic channels
> IFR Jean Roche - Mediterranee University
> Marseille, France

--
Alison J. North, Ph.D.,
Assistant Research Professor and
Director of the Bio-Imaging Resource Center,
The Rockefeller University,
1230 York Avenue,
New York,
NY 10065.
Tel: office ++ 212 327 7488
Tel: lab    ++ 212 327 7486
Fax:        ++ 212 327 7489
Ramshesh, Venkat K Ramshesh, Venkat K
Reply | Threaded
Open this post in threaded view
♦
♦
|

Second Charleston Workshop on LIGHT MICROSCOPY FOR THE BIOSCIENCES (LMB)

In reply to this post by Claudia Campos
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Second Charleston Workshop on
LIGHT MICROSCOPY FOR THE BIOSCIENCES (LMB)
Medical University of South Carolina
May 18-23, 2008

The Second Charleston Workshop on LIGHT MICROSCOPY FOR THE BIOSCIENCES
(LMB) Workshop provides a solid introduction to the concepts and
practical applications of light microscopy relevant to modern cell and
molecular biology. Students will have opportunities for extensive
hands-on experience with state-of-the-art equipment for optical imaging,
digital image processing, and fluorescence and confocal/multiphoton
microscopy guided by experienced academic and commercial faculty.
Lectures and laboratory exercises will include: optics of image
formation; microscope alignment; phase contrast and differential
interference contrast microscopy; video and digital cameras; contrast
enhancement by analog and digital image processing; principles of
fluorescence and fluorescence microscopy; ion imaging and fluorescent
probes, including green fluorescent protein; fluorescence resonance
energy transfer; and laser scanning confocal and multiphoton microscopy.
A commercial faculty representing leading microscope manufacturers will
make available for student use the latest and most advanced
instrumentation for light microscopy, image detection and computerized
image analysis. The LMB Workshop is designed for doctoral level
scientists, advanced pre-doctoral students and high level technical
personnel. No prior experience with microscopy is required. All students
will benefit from in-depth interaction with instructors. Students are
encouraged to bring their own specimens for analysis.

Tuition: $750.00
Application Deadline: April 1, 2008
Principal Instructors:
John J. Lemasters, M.D., Ph.D., Organizer
P. Darwin Bell, Ph.D.
Margaret Kelly, Ph.D.
Peter Komlosi, Ph.D.
Anna-Liisa Nieminen, Ph.D.
Venkat Ramshesh, Ph.D.

To apply, send a curriculum vita and a brief letter describing your
research interests and reasons for enrolling. Because the course is
expected to be oversubscribed, applicants should inquire as soon as
possible. Please indicate your complete mailing address, telephone/fax
number and email address. Full consideration will be given to
applications received by April 1, 2008.

For further information or to apply, contact:
Venkat K Ramshesh
Medical University of South Carolina
Center for Cell Death, Injury and Regeneration and Hollings Cancer Center
280 Calhoun Street, PO Box 250140
Charleston, SC 29425
Telephone (843) 792- 3530, FAX (843) 792-1617
E-mail: [hidden email]
Bjornsson, Chris Bjornsson, Chris
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Secondary antibody for 405nm laser line ?

In reply to this post by lechristophe
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Hi Christophe,

I've had great luck with Alexa405, and highly recommend using this instead of Hoechst, which has a broad enough emission it can be problematic even for spectral imaging and linear unmixing. If you still want a channel for nuclei, CyQuant works great. I'm imaging a sample with Alexa405 right now actually; the only caution is that it seems to photobleach more quickly than my other 4 labels; but I'm trying to burn through 80-micron thick tissue slices, so this may not be typical...

Cheers,
Chris

Chris Bjornsson
Director, Microscopy & Imaging Research Core
Center for Biotechnology and Interdisciplinary Studies, Room 2149
Rensselaer Polytechnic Institute
110 8th St., Troy, NY 12180
ph. 518-276-3456



-----Original Message-----
From: Confocal Microscopy List on behalf of Christophe Leterrier
Sent: Wed 3/5/2008 10:45 AM
To: [hidden email]
Subject: Secondary antibody for 405nm laser line ?
 
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Dear list,

I would like to use all four lines of our new confocal for immunolabeling
(405,488,532,633). For the moment I use Hoechst/Alexa488/Alexa555/Cy5, but I
would like to have a fourth antibody labeling using the 405 channel instead
of Hoechst. What secondary fluorescent antibody would you recommend ? Is
there an Alexa405 or something similar that would be bright and stable ?

Thanks a lot,

Christophe Leterrier

Postdoc
INSERM UMR641 Neurobiology of ionic channels
IFR Jean Roche - Mediterranee University
Marseille, France
simon walker (BI) simon walker (BI)
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Secondary antibody for 405nm laser line ?

In reply to this post by lechristophe
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

We have used Alexa Fluor 405 (Invitrogen) with limited success.  Our experience of it is that it’s reasonably bright, but bleaches very quickly both under epi-illumination and with confocal lsm.  Mounting with an antifade seems to help (we’ve used Vectasheild), but it still bleaches very quickly relative to, for example, AF488.  If someone has found a bright and stable 405-excitable fluorescent dye I would be very interested to know.  Perhaps Qdots maybe a possible alternative?

Simon

Babraham Institute

Cambridge

UK

 


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christophe Leterrier
Sent: 05 March 2008 15:46
To: [hidden email]
Subject: Secondary antibody for 405nm laser line ?

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear list,

I would like to use all four lines of our new confocal for immunolabeling (405,488,532,633). For the moment I use Hoechst/Alexa488/Alexa555/Cy5, but I would like to have a fourth antibody labeling using the 405 channel instead of Hoechst. What secondary fluorescent antibody would you recommend ? Is there an Alexa405 or something similar that would be bright and stable ?

Thanks a lot,

Christophe Leterrier

Postdoc
INSERM UMR641 Neurobiology of ionic channels
IFR Jean Roche - Mediterranee University
Marseille, France

Schebique Schebique
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Secondary antibody for 405nm laser line ?

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

What about Lucifer yellow? It is excitable at 405 but emission is in yellow. With AOBS and sequential scanning mode it should be theoretically distinguishable from others green, yellow and red dyes. But I have never work with that in this way, so I don’t know exactly.

 

Sincereley Ondrej

--------------------------------------------

Ondřej Šebesta

Faculty of Science, Charles University in Prague

Vinicna 7

128 44 Prague

Czech Republic

 

Phone: +420 2 2195 1943

e-mail: [hidden email]

 

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of simon walker (BI)
Sent: Wednesday, March 05, 2008 5:17 PM
To: [hidden email]
Subject: Re: Secondary antibody for 405nm laser line ?

 

We have used Alexa Fluor 405 (Invitrogen) with limited success.  Our experience of it is that it’s reasonably bright, but bleaches very quickly both under epi-illumination and with confocal lsm.  Mounting with an antifade seems to help (we’ve used Vectasheild), but it still bleaches very quickly relative to, for example, AF488.  If someone has found a bright and stable 405-excitable fluorescent dye I would be very interested to know.  Perhaps Qdots maybe a possible alternative?

Simon

Babraham Institute

Cambridge

UK

 


From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christophe Leterrier
Sent: 05 March 2008 15:46
To: [hidden email]
Subject: Secondary antibody for 405nm laser line ?

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear list,

I would like to use all four lines of our new confocal for immunolabeling (405,488,532,633). For the moment I use Hoechst/Alexa488/Alexa555/Cy5, but I would like to have a fourth antibody labeling using the 405 channel instead of Hoechst. What secondary fluorescent antibody would you recommend ? Is there an Alexa405 or something similar that would be bright and stable ?

Thanks a lot,

Christophe Leterrier

Postdoc
INSERM UMR641 Neurobiology of ionic channels
IFR Jean Roche - Mediterranee University
Marseille, France

Asson-Batres, Mary Ann Asson-Batres, Mary Ann
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Secondary antibody for 405nm laser line ?

In reply to this post by simon walker (BI)
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Out of curiosity, what is the problem you are having with Hoechst?  Or are you looking for a 405 marker label to improve specificity?  
 
Mary Ann
Tennessee State University
Nashville, TN  USA

________________________________

From: Confocal Microscopy List on behalf of simon walker (BI)
Sent: Wed 3/5/2008 10:17 AM
To: [hidden email]
Subject: Re: Secondary antibody for 405nm laser line ?


Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal 

We have used Alexa Fluor 405 (Invitrogen) with limited success.  Our experience of it is that it's reasonably bright, but bleaches very quickly both under epi-illumination and with confocal lsm.  Mounting with an antifade seems to help (we've used Vectasheild), but it still bleaches very quickly relative to, for example, AF488.  If someone has found a bright and stable 405-excitable fluorescent dye I would be very interested to know.  Perhaps Qdots maybe a possible alternative?

Simon

Babraham Institute

Cambridge

UK

 

________________________________

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christophe Leterrier
Sent: 05 March 2008 15:46
To: [hidden email]
Subject: Secondary antibody for 405nm laser line ?

 

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear list,

I would like to use all four lines of our new confocal for immunolabeling (405,488,532,633). For the moment I use Hoechst/Alexa488/Alexa555/Cy5, but I would like to have a fourth antibody labeling using the 405 channel instead of Hoechst. What secondary fluorescent antibody would you recommend ? Is there an Alexa405 or something similar that would be bright and stable ?

Thanks a lot,

Christophe Leterrier

Postdoc
INSERM UMR641 Neurobiology of ionic channels
IFR Jean Roche - Mediterranee University
Marseille, France
lechristophe lechristophe
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Secondary antibody for 405nm laser line ?

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal I have non problem with Hoechst, I use it for DNA/Nucleus labeling and it works very well. It's just that I would like to use the 405 channel to label something else (not the DNA) with an antibody, so I'm looking for a fluorescent secondary antibody suitable for 405nm exitation.

Christophe

On Wed, Mar 5, 2008 at 5:35 PM, Asson-Batres, Mary <[hidden email]> wrote:
Out of curiosity, what is the problem you are having with Hoechst?  Or are you looking for a 405 marker label to improve specificity?

Mary Ann
Tennessee State University
Nashville, TN  USA

________________________________

From: Confocal Microscopy List on behalf of simon walker (BI)
Sent: Wed 3/5/2008 10:17 AM
Subject: Re: Secondary antibody for 405nm laser line ?


Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

We have used Alexa Fluor 405 (Invitrogen) with limited success.  Our experience of it is that it's reasonably bright, but bleaches very quickly both under epi-illumination and with confocal lsm.  Mounting with an antifade seems to help (we've used Vectasheild), but it still bleaches very quickly relative to, for example, AF488.  If someone has found a bright and stable 405-excitable fluorescent dye I would be very interested to know.  Perhaps Qdots maybe a possible alternative?

Simon

Babraham Institute

Cambridge

UK



________________________________

From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Christophe Leterrier
Sent: 05 March 2008 15:46
To: [hidden email]
Subject: Secondary antibody for 405nm laser line ?



Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear list,

I would like to use all four lines of our new confocal for immunolabeling (405,488,532,633). For the moment I use Hoechst/Alexa488/Alexa555/Cy5, but I would like to have a fourth antibody labeling using the 405 channel instead of Hoechst. What secondary fluorescent antibody would you recommend ? Is there an Alexa405 or something similar that would be bright and stable ?

Thanks a lot,

Christophe Leterrier

Postdoc
INSERM UMR641 Neurobiology of ionic channels
IFR Jean Roche - Mediterranee University
Marseille, France

Michael Weber-4 Michael Weber-4
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Secondary antibody for 405nm laser line ?

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Christophe,

unfortunately I have to agree that Alexa 405 bleaches rather fast. So
far I am not aware of other commercial secondary antibody for 405 nm.

Michael


Christophe Leterrier wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal I have non
> problem with Hoechst, I use it for DNA/Nucleus labeling and it works
> very well. It's just that I would like to use the 405 channel to label
> something else (not the DNA) with an antibody, so I'm looking for a
> fluorescent secondary antibody suitable for 405nm exitation.
>
> Christophe
>
> On Wed, Mar 5, 2008 at 5:35 PM, Asson-Batres, Mary
> <[hidden email] <mailto:[hidden email]>> wrote:
>
>     Search the CONFOCAL archive at
>     http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>     Out of curiosity, what is the problem you are having with Hoechst?
>      Or are you looking for a 405 marker label to improve specificity?
>
>     Mary Ann
>     Tennessee State University
>     Nashville, TN  USA
>
>     ________________________________
>
>     From: Confocal Microscopy List on behalf of simon walker (BI)
>     Sent: Wed 3/5/2008 10:17 AM
>     To: [hidden email] <mailto:[hidden email]>
>     Subject: Re: Secondary antibody for 405nm laser line ?
>
>
>     Search the CONFOCAL archive at
>     http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>     We have used Alexa Fluor 405 (Invitrogen) with limited success.  Our
>     experience of it is that it's reasonably bright, but bleaches very
>     quickly both under epi-illumination and with confocal lsm.  Mounting
>     with an antifade seems to help (we've used Vectasheild), but it
>     still bleaches very quickly relative to, for example, AF488.  If
>     someone has found a bright and stable 405-excitable fluorescent dye
>     I would be very interested to know.  Perhaps Qdots maybe a possible
>     alternative?
>
>     Simon
>
>     Babraham Institute
>
>     Cambridge
>
>     UK
>
>
>
>     ________________________________
>
>     From: Confocal Microscopy List [mailto:[hidden email]
>     <mailto:[hidden email]>] On Behalf Of Christophe
>     Leterrier
>     Sent: 05 March 2008 15:46
>     To: [hidden email] <mailto:[hidden email]>
>     Subject: Secondary antibody for 405nm laser line ?
>
>
>
>     Search the CONFOCAL archive at
>     http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear list,
>
>     I would like to use all four lines of our new confocal for
>     immunolabeling (405,488,532,633). For the moment I use
>     Hoechst/Alexa488/Alexa555/Cy5, but I would like to have a fourth
>     antibody labeling using the 405 channel instead of Hoechst. What
>     secondary fluorescent antibody would you recommend ? Is there an
>     Alexa405 or something similar that would be bright and stable ?
>
>     Thanks a lot,
>
>     Christophe Leterrier
>
>     Postdoc
>     INSERM UMR641 Neurobiology of ionic channels
>     IFR Jean Roche - Mediterranee University
>     Marseille, France
M. van de corput M. van de corput
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Secondary antibody for 405nm laser line ?

In reply to this post by simon walker (BI)
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

For those not knowing, alexa fluor dyes are reported to bleach more
quickly when embedded in vectashield. Invitrogen has ProLong (gold)
antifading embedding medium that is more suitable for all alexa dyes. The
bleaching is much less and the dyes are more stable over time. I have
re-used slides that were almost 1 year old that were stored at 4oC. The
dyes were less bright and bleached more quickly but for a year old slide I
was surprised it was still that good.
Pro-long gold needs curing overnight and at least 4 days of absolute
curing. Only then the refractive index gets to 1.45. It is not wise to
immediatly go look at the slides when just embedded. The signals will be
weak and blurry.

As for blue dyes. My experience with Alexa Fluor 350 is not that good. It
works but for small targets the dyes are too weak as compared to Alexa488
or 594. For bigger/abundant targets it's fine.

Mariette
Mariette

____________________________________________

Dr. M.P.C. Kemner-van de Corput
____________________________________________

MGC - Dept. of Cell Biology & Genetics
Erasmus Medical Center
Dr. Molewaterplein 50, 3015 GE Rotterdam
POB 2040, 3000 CA Rotterdam, The Netherlands

Delivery adres:
Erasmus MC, Westzeedijk 353, 3015 AA Rotterdam,
The Netherlands

Office: H-Ee751; tel: +31 10 704.3949
Lab: H-Ee710; tel lab: +31 10 704.3315
tel secr: +31 10 704.3169
____________________________________________

http://www2.eur.nl/fgg/ch1/cellbiology/
http://www.thesis.kemner.biz/
____________________________________________

Op Wo, 5 maart, 2008 5:17 pm, schreef simon walker (BI):

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> We have used Alexa Fluor 405 (Invitrogen) with limited success.  Our
> experience of it is that it's reasonably bright, but bleaches very
> quickly both under epi-illumination and with confocal lsm.  Mounting
> with an antifade seems to help (we've used Vectasheild), but it still
> bleaches very quickly relative to, for example, AF488.  If someone has
> found a bright and stable 405-excitable fluorescent dye I would be very
> interested to know.  Perhaps Qdots maybe a possible alternative?
>
> Simon
>
> Babraham Institute
>
> Cambridge
>
> UK
>
>
>
> ________________________________
>
> From: Confocal Microscopy List [mailto:[hidden email]] On
> Behalf Of Christophe Leterrier
> Sent: 05 March 2008 15:46
> To: [hidden email]
> Subject: Secondary antibody for 405nm laser line ?
>
>
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear list,
>
> I would like to use all four lines of our new confocal for
> immunolabeling (405,488,532,633). For the moment I use
> Hoechst/Alexa488/Alexa555/Cy5, but I would like to have a fourth
> antibody labeling using the 405 channel instead of Hoechst. What
> secondary fluorescent antibody would you recommend ? Is there an
> Alexa405 or something similar that would be bright and stable ?
>
> Thanks a lot,
>
> Christophe Leterrier
>
> Postdoc
> INSERM UMR641 Neurobiology of ionic channels
> IFR Jean Roche - Mediterranee University
> Marseille, France
>
>
simon walker (BI) simon walker (BI)
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Secondary antibody for 405nm laser line ?

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

It is not correct to say that ProLong Gold outperforms Vectasheild with
alexa dyes.  Invitrogen's own data show that Vectasheild outperforms
ProLong Gold with Alexa dyes excitable at wavelengths of 488 nm and
less.  It is only the Alexa dyes that are excitable in the green and
particularly the red part of the spectrum that significantly benefit
from ProLong.
Simon

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Mariette P.C. Kemner - van de Corput
Sent: 06 March 2008 12:21
To: [hidden email]
Subject: Re: Secondary antibody for 405nm laser line ?

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

For those not knowing, alexa fluor dyes are reported to bleach more
quickly when embedded in vectashield. Invitrogen has ProLong (gold)
antifading embedding medium that is more suitable for all alexa dyes.
The
bleaching is much less and the dyes are more stable over time. I have
re-used slides that were almost 1 year old that were stored at 4oC. The
dyes were less bright and bleached more quickly but for a year old slide
I
was surprised it was still that good.
Pro-long gold needs curing overnight and at least 4 days of absolute
curing. Only then the refractive index gets to 1.45. It is not wise to
immediatly go look at the slides when just embedded. The signals will be
weak and blurry.

As for blue dyes. My experience with Alexa Fluor 350 is not that good.
It
works but for small targets the dyes are too weak as compared to
Alexa488
or 594. For bigger/abundant targets it's fine.

Mariette
Mariette

____________________________________________

Dr. M.P.C. Kemner-van de Corput
____________________________________________

MGC - Dept. of Cell Biology & Genetics
Erasmus Medical Center
Dr. Molewaterplein 50, 3015 GE Rotterdam
POB 2040, 3000 CA Rotterdam, The Netherlands

Delivery adres:
Erasmus MC, Westzeedijk 353, 3015 AA Rotterdam,
The Netherlands

Office: H-Ee751; tel: +31 10 704.3949
Lab: H-Ee710; tel lab: +31 10 704.3315
tel secr: +31 10 704.3169
____________________________________________

http://www2.eur.nl/fgg/ch1/cellbiology/
http://www.thesis.kemner.biz/
____________________________________________

Op Wo, 5 maart, 2008 5:17 pm, schreef simon walker (BI):
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> We have used Alexa Fluor 405 (Invitrogen) with limited success.  Our
> experience of it is that it's reasonably bright, but bleaches very
> quickly both under epi-illumination and with confocal lsm.  Mounting
> with an antifade seems to help (we've used Vectasheild), but it still
> bleaches very quickly relative to, for example, AF488.  If someone has
> found a bright and stable 405-excitable fluorescent dye I would be
very

> interested to know.  Perhaps Qdots maybe a possible alternative?
>
> Simon
>
> Babraham Institute
>
> Cambridge
>
> UK
>
>
>
> ________________________________
>
> From: Confocal Microscopy List [mailto:[hidden email]]
On

> Behalf Of Christophe Leterrier
> Sent: 05 March 2008 15:46
> To: [hidden email]
> Subject: Secondary antibody for 405nm laser line ?
>
>
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear list,
>
> I would like to use all four lines of our new confocal for
> immunolabeling (405,488,532,633). For the moment I use
> Hoechst/Alexa488/Alexa555/Cy5, but I would like to have a fourth
> antibody labeling using the 405 channel instead of Hoechst. What
> secondary fluorescent antibody would you recommend ? Is there an
> Alexa405 or something similar that would be bright and stable ?
>
> Thanks a lot,
>
> Christophe Leterrier
>
> Postdoc
> INSERM UMR641 Neurobiology of ionic channels
> IFR Jean Roche - Mediterranee University
> Marseille, France
>
>
Benoit Labarthe Benoit Labarthe
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Secondary antibody for 405nm laser line ?

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal hello,

we used AnnexinV conjugated to Pacific Blue, on a spinning disk confocal with reasonnable bleaching.

benoit

On Thu, Mar 6, 2008 at 7:30 AM, simon walker (BI) <[hidden email]> wrote:
It is not correct to say that ProLong Gold outperforms Vectasheild with
alexa dyes.  Invitrogen's own data show that Vectasheild outperforms
ProLong Gold with Alexa dyes excitable at wavelengths of 488 nm and
less.  It is only the Alexa dyes that are excitable in the green and
particularly the red part of the spectrum that significantly benefit
from ProLong.
Simon

-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]] On
Behalf Of Mariette P.C. Kemner - van de Corput
Sent: 06 March 2008 12:21
To: [hidden email]
Subject: Re: Secondary antibody for 405nm laser line ?

Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

For those not knowing, alexa fluor dyes are reported to bleach more
quickly when embedded in vectashield. Invitrogen has ProLong (gold)
antifading embedding medium that is more suitable for all alexa dyes.
The
bleaching is much less and the dyes are more stable over time. I have
re-used slides that were almost 1 year old that were stored at 4oC. The
dyes were less bright and bleached more quickly but for a year old slide
I
was surprised it was still that good.
Pro-long gold needs curing overnight and at least 4 days of absolute
curing. Only then the refractive index gets to 1.45. It is not wise to
immediatly go look at the slides when just embedded. The signals will be
weak and blurry.

As for blue dyes. My experience with Alexa Fluor 350 is not that good.
It
works but for small targets the dyes are too weak as compared to
Alexa488
or 594. For bigger/abundant targets it's fine.

Mariette
Mariette

____________________________________________

Dr. M.P.C. Kemner-van de Corput
____________________________________________

MGC - Dept. of Cell Biology & Genetics
Erasmus Medical Center
Dr. Molewaterplein 50, 3015 GE Rotterdam
POB 2040, 3000 CA Rotterdam, The Netherlands

Delivery adres:
Erasmus MC, Westzeedijk 353, 3015 AA Rotterdam,
The Netherlands

Office: H-Ee751; tel: +31 10 704.3949
Lab: H-Ee710; tel lab: +31 10 704.3315
tel secr: +31 10 704.3169
____________________________________________

http://www2.eur.nl/fgg/ch1/cellbiology/
http://www.thesis.kemner.biz/
____________________________________________

Op Wo, 5 maart, 2008 5:17 pm, schreef simon walker (BI):
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> We have used Alexa Fluor 405 (Invitrogen) with limited success.  Our
> experience of it is that it's reasonably bright, but bleaches very
> quickly both under epi-illumination and with confocal lsm.  Mounting
> with an antifade seems to help (we've used Vectasheild), but it still
> bleaches very quickly relative to, for example, AF488.  If someone has
> found a bright and stable 405-excitable fluorescent dye I would be
very
> interested to know.  Perhaps Qdots maybe a possible alternative?
>
> Simon
>
> Babraham Institute
>
> Cambridge
>
> UK
>
>
>
> ________________________________
>
> From: Confocal Microscopy List [mailto:[hidden email]]
On
> Behalf Of Christophe Leterrier
> Sent: 05 March 2008 15:46
> To: [hidden email]
> Subject: Secondary antibody for 405nm laser line ?
>
>
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear list,
>
> I would like to use all four lines of our new confocal for
> immunolabeling (405,488,532,633). For the moment I use
> Hoechst/Alexa488/Alexa555/Cy5, but I would like to have a fourth
> antibody labeling using the 405 channel instead of Hoechst. What
> secondary fluorescent antibody would you recommend ? Is there an
> Alexa405 or something similar that would be bright and stable ?
>
> Thanks a lot,
>
> Christophe Leterrier
>
> Postdoc
> INSERM UMR641 Neurobiology of ionic channels
> IFR Jean Roche - Mediterranee University
> Marseille, France
>
>



--
Benoît Labarthe, PharmD
Institut de Cardiologie de Montréal
Montreal Heart Institute
Centre de Recherche, Laboratoire du Dr.Théroux
5000 Bélanger Est
Montréal Québec Canada H1T 1C8
Tel (514)-376 3330 ext 3017
Fax (514)- 376 1076
Bjornsson, Chris Bjornsson, Chris
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Secondary antibody for 405nm laser line ?

In reply to this post by M. van de corput
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I use ProLong and a handful of different Alexa dyes every time; the A405 still bleaches faster than the others, unfortunately!

Cheers,
Chris

Chris Bjornsson
Director, Microscopy & Imaging Research Core
Center for Biotechnology and Interdisciplinary Studies, Room 2149
Rensselaer Polytechnic Institute
110 8th St., Troy, NY 12180
ph. 518-276-3456



-----Original Message-----
From: Confocal Microscopy List on behalf of Mariette P.C. Kemner - van de Corput
Sent: Thu 3/6/2008 7:21 AM
To: [hidden email]
Subject: Re: Secondary antibody for 405nm laser line ?
 
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

For those not knowing, alexa fluor dyes are reported to bleach more
quickly when embedded in vectashield. Invitrogen has ProLong (gold)
antifading embedding medium that is more suitable for all alexa dyes. The
bleaching is much less and the dyes are more stable over time. I have
re-used slides that were almost 1 year old that were stored at 4oC. The
dyes were less bright and bleached more quickly but for a year old slide I
was surprised it was still that good.
Pro-long gold needs curing overnight and at least 4 days of absolute
curing. Only then the refractive index gets to 1.45. It is not wise to
immediatly go look at the slides when just embedded. The signals will be
weak and blurry.

As for blue dyes. My experience with Alexa Fluor 350 is not that good. It
works but for small targets the dyes are too weak as compared to Alexa488
or 594. For bigger/abundant targets it's fine.

Mariette
Mariette

____________________________________________

Dr. M.P.C. Kemner-van de Corput
____________________________________________

MGC - Dept. of Cell Biology & Genetics
Erasmus Medical Center
Dr. Molewaterplein 50, 3015 GE Rotterdam
POB 2040, 3000 CA Rotterdam, The Netherlands

Delivery adres:
Erasmus MC, Westzeedijk 353, 3015 AA Rotterdam,
The Netherlands

Office: H-Ee751; tel: +31 10 704.3949
Lab: H-Ee710; tel lab: +31 10 704.3315
tel secr: +31 10 704.3169
____________________________________________

http://www2.eur.nl/fgg/ch1/cellbiology/
http://www.thesis.kemner.biz/
____________________________________________

Op Wo, 5 maart, 2008 5:17 pm, schreef simon walker (BI):

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> We have used Alexa Fluor 405 (Invitrogen) with limited success.  Our
> experience of it is that it's reasonably bright, but bleaches very
> quickly both under epi-illumination and with confocal lsm.  Mounting
> with an antifade seems to help (we've used Vectasheild), but it still
> bleaches very quickly relative to, for example, AF488.  If someone has
> found a bright and stable 405-excitable fluorescent dye I would be very
> interested to know.  Perhaps Qdots maybe a possible alternative?
>
> Simon
>
> Babraham Institute
>
> Cambridge
>
> UK
>
>
>
> ________________________________
>
> From: Confocal Microscopy List [mailto:[hidden email]] On
> Behalf Of Christophe Leterrier
> Sent: 05 March 2008 15:46
> To: [hidden email]
> Subject: Secondary antibody for 405nm laser line ?
>
>
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear list,
>
> I would like to use all four lines of our new confocal for
> immunolabeling (405,488,532,633). For the moment I use
> Hoechst/Alexa488/Alexa555/Cy5, but I would like to have a fourth
> antibody labeling using the 405 channel instead of Hoechst. What
> secondary fluorescent antibody would you recommend ? Is there an
> Alexa405 or something similar that would be bright and stable ?
>
> Thanks a lot,
>
> Christophe Leterrier
>
> Postdoc
> INSERM UMR641 Neurobiology of ionic channels
> IFR Jean Roche - Mediterranee University
> Marseille, France
>
>
Jerry Sedgewick-2 Jerry Sedgewick-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Tsunami for sale

In reply to this post by Benoit Labarthe
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

We have aTi:Sapphire laser system we are interested in selling.  This is
the mode-locked Ti:Sapphire laser (Tsunami 3960, Spectra Physics Inc.,
Mountain View, CA) pumped by a high-power (5 W, 532 nm) laser (Millennia
V, Spectra Physics Inc.). It comes with a spectrometer.  One of the
diode lasers used in the Millenia is not operational: if you're
interested, then we'd have that repaired.

If this is of interest to you, please make an offer with the idea that
the Millennia is repaired and in prime operating condition.

Send email to:  Jerry Sedgewick: [hidden email].

Thanks!

Jerry Sedgewick
Program Director, Biomedical Image Processing Lab
University of Minnesota

ÐÏࡱá




--- http://USFamily.Net/dialup.html - $8.25/mo! -- http://www.usfamily.net/dsl.html - $19.99/mo! ---
Lenaldo Rocha Lenaldo Rocha
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Secondary antibody for 405nm laser line ?

In reply to this post by lechristophe
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Christophe

I've read that cascade blue excited with 405 laser is as bright as
Alexa488, that might be a good option. I don't have personal experience
with it so can't tell if it fades faster than any other dye.

Lenaldo


Christophe Leterrier wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Dear list,
>
> I would like to use all four lines of our new confocal for
> immunolabeling (405,488,532,633). For the moment I use
> Hoechst/Alexa488/Alexa555/Cy5, but I would like to have a fourth
> antibody labeling using the 405 channel instead of Hoechst. What
> secondary fluorescent antibody would you recommend ? Is there an
> Alexa405 or something similar that would be bright and stable ?
>
> Thanks a lot,
>
> Christophe Leterrier
>
> Postdoc
> INSERM UMR641 Neurobiology of ionic channels
> IFR Jean Roche - Mediterranee University
> Marseille, France
Junya HIROI Junya HIROI
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Secondary antibody for 405nm laser line ?

In reply to this post by lechristophe
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

Christophe,

I was also looking for a fluorescent dye for 405 nm excitation, and I compared Alexa 405
with Pacific Blue.  Both dyes bleached to some extent with ProlongGold, but Pacific Blue was
absolutely brighter than Alexa 405, when compared at the same dilution.  So, I recommend
Pacific Blue, although Molecular Probes sells Pacific Blue for flow cytometry.

I found another problem on Alexa 405: it nonspecifically binds to nuclei !  I asked Molecular
Probes/Invitrogen about this problem, and they recommended me to use their blocking
solution, Image-iT FX Signal Enhancer.  It actually worked well, but Pacific Blue does not have
such a problem, so that is another reason I like Pacific Blue.

Junya

--
_____________________________________
œAˆä€–ç
¹ƒ}ƒŠƒAƒ“ƒiˆã‰È‘åŠw ‰ð–UŠw‹³Žº
§216-8511 ìèŽs‹{‘O‹æ›¶2-16-1
tel: 044-977-8111 “àü3629
fax: 044-976-7083
e-mail: [hidden email]
_____________________________________

--
Fredrik Wermeling Fredrik Wermeling
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Secondary antibody for 405nm laser line ?

In reply to this post by lechristophe
Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

I´ve also used Pacific blue with good result. The target for the conjugated antibody I used has a very high expression which I guess makes it more easy to get a good signal. /fredrik

 

----- Original Message -----
From: Christophe Leterrier <[hidden email]>
Date: Wednesday, March 5, 2008 10:58 am
Subject: Secondary antibody for 405nm laser line ?
To: [hidden email]


> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Dear list,
>
> I would like to use all four lines of our new confocal for
> immunolabeling(405,488,532,633). For the moment I use
> Hoechst/Alexa488/Alexa555/Cy5, but I
> would like to have a fourth antibody labeling using the 405
> channel instead
> of Hoechst. What secondary fluorescent antibody would you
> recommend ? Is
> there an Alexa405 or something similar that would be bright and
> stable ?
>
> Thanks a lot,
>
> Christophe Leterrier
>
> Postdoc
> INSERM UMR641 Neurobiology of ionic channels
> IFR Jean Roche - Mediterranee University
> Marseille, France
>

lechristophe lechristophe
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Secondary antibody for 405nm laser line ?

Search the CONFOCAL archive at http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal Thank you all for your responses and recommandation. I ordered goat anti-rabbit Alexa405 antibody and will keep you informed about its performances. I'd be very interested in an anti-chicken secondary antibody that can be excited at 405 nm (because I'm using a chicken antibody to stain an abundant structural marker in my cells, keeping the rabbit and mouse antibodies for more exotic targets). Unfortunately Invitrogen/Molecular Probes does not sell anti-chicken secondaries conjuguated to Alexa 405 or Cascade Blue :(

Christophe Leterrier
Glen MacDonald-2 Glen MacDonald-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Secondary antibody for 405nm laser line ?

In reply to this post by Junya HIROI
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

All of the Alexa fluor conjugates have a tendency to stick non-
specifically via charge interactions of the anionic dye moiety.  the  
Image-iT fx does helps greatly, especially in thick samples.

In regard to using the blue fluorophores for the more prominent  
targets in the tissue, the human visual system is only weakly  
responsive to blue wavelengths and lacks blue-sensitive cones in the  
fovea, the region of greatest visual acuity.  In general, if the  
smallest or rarest bits get the green, red or gray channel, they are  
noticeable with less histogram manipulation than if the smallest bits  
are viewed in the blue channel.

    Regards,
Glen



Glen MacDonald
Core for Communication Research
Virginia Merrill Bloedel Hearing Research Center
Box 357923
University of Washington
Seattle, WA 98195-7923  USA
(206) 616-4156
[hidden email]

************************************************************************
******
The box said "Requires WindowsXP or better", so I bought a Macintosh.
************************************************************************
******


On Mar 6, 2008, at 7:09 AM, Junya HIROI wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Christophe,
>
> I was also looking for a fluorescent dye for 405 nm excitation, and  
> I compared Alexa 405
> with Pacific Blue.  Both dyes bleached to some extent with  
> ProlongGold, but Pacific Blue was
> absolutely brighter than Alexa 405, when compared at the same  
> dilution.  So, I recommend
> Pacific Blue, although Molecular Probes sells Pacific Blue for flow  
> cytometry.
>
> I found another problem on Alexa 405: it nonspecifically binds to  
> nuclei !  I asked Molecular
> Probes/Invitrogen about this problem, and they recommended me to  
> use their blocking
> solution, Image-iT FX Signal Enhancer.  It actually worked well,  
> but Pacific Blue does not have
> such a problem, so that is another reason I like Pacific Blue.
>
> Junya
>
> --
> _____________________________________
> œAˆä€–ç
> ¹ƒ}ƒŠƒAƒ“ƒiˆã‰È‘åŠw ‰ð–UŠw‹³Žº
> §216-8511 ìèŽs‹{‘O‹æ›¶2-16-1
> tel: 044-977-8111 “àü3629
> fax: 044-976-7083
> e-mail: [hidden email]
> _____________________________________
>
> --