Jacqueline Ross |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear All, I'm currently trialling 2 silicone oil immersion lenses from Olympus (30x/1.05NA and 60x/1.3NA) for live cell imaging on our Olympus FV1000 confocal, which is on an upright fixed stage microscope. So far, they both seem excellent and offer the higher level of resolution I'm looking for. I am particularly interested in how these lenses compare with high NA water-immersion objective lenses. At the moment, we have long working distance water-immersion lenses or standard oil immersion so I don't have a high NA water-immersion objective lens to do my own comparison. If anyone has personal experience with both types of lenses, I would be keen to hear their view. I'm particularly interested in how the 60x/1.2NA water-immersion lens compares to the silicone oil immersion 60x/1.3NA, since I would like to buy one or the other. Kind regards, Jacqui Jacqueline Ross Biomedical Imaging Microscopist Biomedical Imaging Research Unit School of Medical Sciences Faculty of Medical & Health Sciences The University of Auckland Private Bag 92019 Auckland, NEW ZEALAND Tel: 64 9 373 7599 Ext 87438 Fax: 64 9 373 7484 http://www.fmhs.auckland.ac.nz/sms/biru/ |
|
Pascal Weber |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I have no experience with these two objetives. A long time ago but I had done this compraison between two objectives same magnification water and oil. The conclusion was that oil was higher yielding than water and had a better resolution for the water immersion. Just one question how do you proceed to do cells with an upright microscope and immersion objectives ? In my previous lab i do this : on an inverted microscope (Zeiss) i use the water 20X LD obj (75mm) to get a wilde FOV with the best resolution. |
|
Stefan Sokoll |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello, I am doing single particle tracking using an upright confocal spinning disk setup. The setup basically consists of an Olympus TIRF100x oil objective, an Olympus microscope and on top a Yokogawa spinning disk CSU-X1 with an Andor EMCCD camera. Analyzing the 3D PSF of latex beats I discover that the PSF is not symmetrical in xy direction, instead the x expansion is roughly 50nm larger than in y direction. I wonder what might be the main cause for this effect? Tilted cover slip, adjustment of the devices, the CSU itself or ...? So far I just put the devices on top of each other but due to the upright setup I cannot check the symmetry of the laser beam e.g. over long distance at a wall. Any ideas on how to best start calibrating the setup are very welcome. Thanks in advance! Best, Stefan -- --------- Dipl. Ing.-Inf. Stefan Sokoll Lab. Molecular Physiology Dept. Neurochemistry & Molecular Biology Leibniz Institute for Neurobiology (LIN) Brenneckestrasse 6 39118 Magdeburg Germany Mail: [hidden email] Tel.: +49 - 391 - 626393171 |
|
George McNamara |
|
|
In reply to this post by Jacqueline Ross
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Jacqui, The Olympus press release, Olympus has introduced new 30x and 60x silicone oil objectives, its latest innovation in silicone-immersion microscope optics. The new UPLSAPO 30x and UPLSAPO 60x objectives can markedly improve optical performance for live cell confocal, widefield fluorescence, multiphoton, differential interference contrast (DIC) and other applications. The new silicone oil objectives have distinct advantages over both traditional oil- and water-immersion objectives for deep and long -term imaging. They provide higher resolution and brighter images when imaging within deep samples when compared with oil-immersion objectives, and are useful in long-term imaging experiments where, in the past, water evaporation has been an issue. The new 30x, 1.05 NA, 0.8 mm working distance and 60x, 1.3 NA, 0.3 mm working distance objectives reduce loss of contrast due to spherical aberration when compared to conventional oil objectives. By reducing the mismatch between the refractive index of the specimen and that of the immersion medium into which the objective is dipped, they provide higher resolution and brightness when imaging into thick samples. While water-immersion lenses are traditionally used to reduce refraction index mismatch, they are not practical for longterm, time-lapse imaging because of evaporation and low viscosity. Glycerol-immersion optics have been used as a substitute for water, but glycerol is not ideal because it tends to draw moisture from the air, resulting in changes in refractive index over time. Silicone oil is very stable and does not have these issues, and more closely matches the refractive index (N=1.404) of intracellular components, making these new objectives useful for imaging into cells and also for long-term, time-lapse experiments. The new objectives also feature a correction collar, enabling the user to correct for the spherical aberration that is present when imaging into a specimen beyond the coverslip. Image resolution and contrast, along with fluorescence performance, are optimized and maintained by adjusting this collar. states the RI = 1.404 better matches "intracellular components". So, if you are imaging entirely through homogeneous cells, the silicone lens should win, no matter how far into the specimen you image in Z. If you are imaging a cell-sparse specimen, that is, imaging mostly through saline or saline + some percent serum, an RI=1.333 water immersion lens will likely win at depth -- I hope none of your users just want to image a volume of 50 um of just H2O. The brightness curves with depth should behave similar - in the sense of RI matching should be best - to the Staudt ... Hell Microscopy Res Tech paper's graph. The Olympus p.r. does not mention an interesting (but possibly labor intensive for you to do) solution: increase the R.I. of the cell culture medium to close to 1.404. On page 621 of my mpmicro.doc is a table with R.I'.s of cell components (actually, several tables from different sources, none of which agree perfectly, below is the table from one source ... if anyone has a different version of mpmicro from the web, just search for Widder): Johnsen and Widder (1999) (their table 1) list the following values for different cell components: Cell Compartment Absolute refractive index Reference Cell membrane 1.46-1.60 Quinby-Hunt & Hunt (1988) Cell membrane 1.46-1.54 Meyer (1979) Collagen 1.55 Chapman (1976) Cytoplasm 1.35 Charney & Brackett (1961) Lipids 1.48 Beuthan et al. (1996) Mitochondria 1.40 Beuthan et al. (1996) S. Johnsen, E. A. Widder (1999) The physical basis of transparency in biological tissue: ultrastructure and the minimization of light scattering. J Theor Biol 199:181--198. (ok, maybe picking values from a theoretical journal is not the best possible choice). One upshot of the table above is that cells are not homogeneous. Other sources give different values for "cytoplasm", including: Cytoplasm 1.374 Cytoplasm (range) 1.358-1.374 Cytoplasm (range) 1.353-1.368 Relative dry mass of 11% to 19%, for regions of cytoplasm outside focal and close contacts. I recall calculating that a whole lot of BSA or HSA (for examples) could be used to increase the R.I. to get closer to ~1.4 - however, I don't know what that will do to scattering properties or viscosity. So, I suggest testing brightness, resolution, and performance with different fluorophores, in Z-series. Numerically correct deconvolution of all your confocal data sets would be useful. If possible, measure the thickness of each coverglass (and make sure each is on flat). Sincerely, George p.s. may everyone's light rays be as straight as possible. On 6/28/2012 2:47 AM, Jacqui Ross wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear All, > > I'm currently trialling 2 silicone oil immersion lenses from Olympus (30x/1.05NA and 60x/1.3NA) for live cell imaging on our Olympus FV1000 confocal, which is on an upright fixed stage microscope. So far, they both seem excellent and offer the higher level of resolution I'm looking for. > > I am particularly interested in how these lenses compare with high NA water-immersion objective lenses. At the moment, we have long working distance water-immersion lenses or standard oil immersion so I don't have a high NA water-immersion objective lens to do my own comparison. > > If anyone has personal experience with both types of lenses, I would be keen to hear their view. > > I'm particularly interested in how the 60x/1.2NA water-immersion lens compares to the silicone oil immersion 60x/1.3NA, since I would like to buy one or the other. > > Kind regards, > > Jacqui > > Jacqueline Ross > Biomedical Imaging Microscopist > Biomedical Imaging Research Unit > School of Medical Sciences > Faculty of Medical& Health Sciences > The University of Auckland > Private Bag 92019 > Auckland, NEW ZEALAND > > Tel: 64 9 373 7599 Ext 87438 > Fax: 64 9 373 7484 > > http://www.fmhs.auckland.ac.nz/sms/biru/ > > |
|
Sebastian Rhode |
|
|
In reply to this post by Stefan Sokoll
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Stefan, from my experience so far, there can be a couple of things go wrong. When I found asymmetric PSF, the most common problem were caused by: - the objective itself - Did you check the PSF of those beads using the same objective on different setups? - Do you use a CMount with built-in optics (Aftermagnification) since use a CCD with rather large pixels? - Dichroic - especially the CSU-X1 use very thin (0.5mm) dichroics, which will not be very flat (for this reason we use 5mm substrates for our dicroics) - so this might induce a curvature in one direction and finally lead to astigmatism By the way, what are the values you calculated? By the way, no commercial interests here. Cheers, Sebi Dr. Sebastian Rhode Project Manager Research & Development TILL Photonics GmbH an FEI Company On Thu, 28 Jun 2012 12:08:02 +0200, Stefan Sokoll <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >***** > >Hello, > >I am doing single particle tracking using an upright confocal spinning disk setup. The setup >basically consists of an Olympus TIRF100x oil objective, an Olympus microscope and on top a Yokogawa >spinning disk CSU-X1 with an Andor EMCCD camera. > >Analyzing the 3D PSF of latex beats I discover that the PSF is not symmetrical in xy direction, >instead the x expansion is roughly 50nm larger than in y direction. > >I wonder what might be the main cause for this effect? Tilted cover slip, adjustment of the devices, >the CSU itself or ...? So far I just put the devices on top of each other but due to the upright >setup I cannot check the symmetry of the laser beam e.g. over long distance at a wall. > >Any ideas on how to best start calibrating the setup are very welcome. Thanks in advance! > >Best, >Stefan > >-- >--------- >Dipl. Ing.-Inf. Stefan Sokoll >Lab. Molecular Physiology >Dept. Neurochemistry & Molecular Biology >Leibniz Institute for Neurobiology (LIN) >Brenneckestrasse 6 >39118 Magdeburg >Germany > >Mail: [hidden email] >Tel.: +49 - 391 - 626393171 |
|
Stefan Sokoll |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Sebastian, I am simply calculating the x/y fwhm of each particle. I already used different objectives, still giving the same results, thus it's probably not a matter of the objective. Also we use no aftermagnification. Unfortunately, I don't understand what you mean with using 5mm substrates on the dicroics. Maybe you can explain this in more detail? Thanks! Stefan Sebastian Rhode wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Stefan, > > from my experience so far, there can be a couple of things go wrong. When I > found asymmetric PSF, the most common problem were caused by: > > - the objective itself - Did you check the PSF of those beads using the same > objective on different setups? > > - Do you use a CMount with built-in optics (Aftermagnification) since use a > CCD with rather large pixels? > > - Dichroic - especially the CSU-X1 use very thin (0.5mm) dichroics, which > will not be very flat (for this reason we use 5mm substrates for our > dicroics) - so this might induce a curvature in one direction and finally > lead to astigmatism > > By the way, what are the values you calculated? > > By the way, no commercial interests here. > > Cheers, Sebi > > Dr. Sebastian Rhode > Project Manager Research & Development > > TILL Photonics GmbH > an FEI Company > > > > On Thu, 28 Jun 2012 12:08:02 +0200, Stefan Sokoll <[hidden email]> > wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> ***** >> >> Hello, >> >> I am doing single particle tracking using an upright confocal spinning disk > setup. The setup >> basically consists of an Olympus TIRF100x oil objective, an Olympus > microscope and on top a Yokogawa >> spinning disk CSU-X1 with an Andor EMCCD camera. >> >> Analyzing the 3D PSF of latex beats I discover that the PSF is not > symmetrical in xy direction, >> instead the x expansion is roughly 50nm larger than in y direction. >> >> I wonder what might be the main cause for this effect? Tilted cover slip, > adjustment of the devices, >> the CSU itself or ...? So far I just put the devices on top of each other > but due to the upright >> setup I cannot check the symmetry of the laser beam e.g. over long distance > at a wall. >> Any ideas on how to best start calibrating the setup are very welcome. > Thanks in advance! >> Best, >> Stefan >> >> -- >> --------- >> Dipl. Ing.-Inf. Stefan Sokoll >> Lab. Molecular Physiology >> Dept. Neurochemistry & Molecular Biology >> Leibniz Institute for Neurobiology (LIN) >> Brenneckestrasse 6 >> 39118 Magdeburg >> Germany >> >> Mail: [hidden email] >> Tel.: +49 - 391 - 626393171 -- --------- Dipl. Ing.-Inf. Stefan Sokoll Lab. Molecular Physiology Dept. Neurochemistry & Molecular Biology Leibniz Institute for Neurobiology (LIN) Brenneckestrasse 6 39118 Magdeburg Germany Mail: [hidden email] Tel.: +49 - 391 - 626393171 |
|
John Oreopoulos |
|
|
In reply to this post by Stefan Sokoll
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Stefan, Would you be able to post to the web somewhere a few example images so we can see as well? Maybe even a XZ and XY cross-section reconstructions through a few beads? Is the effect the same all across the field of view? The shape of the PSF can sometimes tell you if there are any specific aberrations in play. The best way to isolate the problem would be to move the objective over to another scope (without a CSU) and try collecting the same data on the same type of sample. What happens if you use a similar objective on the original scope system? Then you could also try removing the CSU on your first scope and connect the camera directly to the upright port, and again collect the data with the objective in question (use the arc lamp for a light source and scope filter cube). This will determine if it's the objective, the scope, or maybe the CSU. Cheers, John Oreopoulos Research Assistant Spectral Applied Research Richmond Hill, Ontario Canada www.spectral.ca On 2012-06-28, at 6:08 AM, Stefan Sokoll wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello, > > I am doing single particle tracking using an upright confocal spinning disk setup. The setup > basically consists of an Olympus TIRF100x oil objective, an Olympus microscope and on top a Yokogawa > spinning disk CSU-X1 with an Andor EMCCD camera. > > Analyzing the 3D PSF of latex beats I discover that the PSF is not symmetrical in xy direction, > instead the x expansion is roughly 50nm larger than in y direction. > > I wonder what might be the main cause for this effect? Tilted cover slip, adjustment of the devices, > the CSU itself or ...? So far I just put the devices on top of each other but due to the upright > setup I cannot check the symmetry of the laser beam e.g. over long distance at a wall. > > Any ideas on how to best start calibrating the setup are very welcome. Thanks in advance! > > Best, > Stefan > > -- > --------- > Dipl. Ing.-Inf. Stefan Sokoll > Lab. Molecular Physiology > Dept. Neurochemistry & Molecular Biology > Leibniz Institute for Neurobiology (LIN) > Brenneckestrasse 6 > 39118 Magdeburg > Germany > > Mail: [hidden email] > Tel.: +49 - 391 - 626393171 |
|
Sebastian Rhode |
|
|
In reply to this post by Stefan Sokoll
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Stefan, I might be wrong in your case, but I have seen quite a lot of systems, where a rather thin dichroic introduces optical problems. Imagine the dichroic is bend a bit due to its mounting along on axis. So this can lead to a curved surface along one axis of the dichroic. But this curvature introduces a focal length along on axis. And finally those kind of things will "mess up" your otherwise perfect PSF. To avoid we use 2mm substrates (TITF, FRAP, WF) inside the microscope or even 5mm substrates for our dichroics inside our spinning disc system. Cheers, Sebi |
|
Stefan Sokoll |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Ok, now I got it. Fortunately, we have different dicroics so I will exchange them and check for changes on the PSF. I analyze the aberrations not directly on the plot of the PSF instead I analyze the changes of the 2D fwhm over z (separated for x and y direction), which e.g. leads to the graphs on the left in the attached image. There is an offset between the x and y fwhm and both curves also seem a little shifted over z. Choosing a particle that is not centered (here I choose one that is on the outer right of the field of view but same y coordinate) changes the offset/shifting and in this example even leads to a perfect match of the curves (graph on the right). Maybe this also points to a specific aberration? I recognized that attaching files is not allowed on that list so I will send the image directly. Best, Stefan Sebastian Rhode wrote: > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Hi Stefan, > > > > I might be wrong in your case, but I have seen quite a lot of systems, where > > a rather thin dichroic introduces optical problems. Imagine the dichroic is > > bend a bit due to its mounting along on axis. So this can lead to a curved > > surface along one axis of the dichroic. But this curvature introduces a > > focal length along on axis. > > > > And finally those kind of things will "mess up" your otherwise perfect PSF. > > To avoid we use 2mm substrates (TITF, FRAP, WF) inside the microscope or > > even 5mm substrates for our dichroics inside our spinning disc system. > > > > Cheers, Sebi > > Stefan, Would you be able to post to the web somewhere a few example images so we can see as well? Maybe even a XZ and XY cross-section reconstructions through a few beads? Is the effect the same all across the field of view? The shape of the PSF can sometimes tell you if there are any specific aberrations in play. The best way to isolate the problem would be to move the objective over to another scope (without a CSU) and try collecting the same data on the same type of sample. What happens if you use a similar objective on the original scope system? Then you could also try removing the CSU on your first scope and connect the camera directly to the upright port, and again collect the data with the objective in question (use the arc lamp for a light source and scope filter cube). This will determine if it's the objective, the scope, or maybe the CSU. Cheers, John Oreopoulos Research Assistant Spectral Applied Research Richmond Hill, Ontario Canada www.spectral.ca -- --------- Dipl. Ing.-Inf. Stefan Sokoll Lab. Molecular Physiology Dept. Neurochemistry & Molecular Biology Leibniz Institute for Neurobiology (LIN) Brenneckestrasse 6 39118 Magdeburg Germany Mail: [hidden email] Tel.: +49 - 391 - 626393171 |
|
Andrew York |
|
|
In reply to this post by Jacqueline Ross
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Sadly I haven't done head-to-head comparisons, but we've been happy using the 60x silicone objective you mention, and we've also been happy with our Olympus 60x 1.2 water objective on a different system. We've used the silicone for single cells, tissue samples, and embryos, and gotten excellent results >50 microns deep. We've used the water objective primarily in single cells 10 microns thick or less. For both objectives I've found the correction collar to be fairly important, and the setting that gives best results is not always what you would expect from the specified coverslip thickness. For the water objective, the tilt of the coverslip is extremely important (as described here: http://www.ncbi.nlm.nih.gov/pubmed/15369482 ). If you really care about high resolution, demo several of each objective, and pick the one that has the best PSF. We've noticed substantial performance variation for supposedly identical objectives. I think you can also ask Olympus for a "PSF-grade" objective. I think this just means they pick out a good one for you. I'm not sure there's enough silicone objectives to be choosy, but it's worth a shot. Neither of our objectives is "PSF-grade", so I can't say how much better they are. On Thu, Jun 28, 2012 at 2:47 AM, Jacqui Ross <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear All, > > I'm currently trialling 2 silicone oil immersion lenses from Olympus > (30x/1.05NA and 60x/1.3NA) for live cell imaging on our Olympus FV1000 > confocal, which is on an upright fixed stage microscope. So far, they both > seem excellent and offer the higher level of resolution I'm looking for. > > I am particularly interested in how these lenses compare with high NA > water-immersion objective lenses. At the moment, we have long working > distance water-immersion lenses or standard oil immersion so I don't have a > high NA water-immersion objective lens to do my own comparison. > > If anyone has personal experience with both types of lenses, I would be > keen to hear their view. > > I'm particularly interested in how the 60x/1.2NA water-immersion lens > compares to the silicone oil immersion 60x/1.3NA, since I would like to buy > one or the other. > > Kind regards, > > Jacqui > > Jacqueline Ross > Biomedical Imaging Microscopist > Biomedical Imaging Research Unit > School of Medical Sciences > Faculty of Medical & Health Sciences > The University of Auckland > Private Bag 92019 > Auckland, NEW ZEALAND > > Tel: 64 9 373 7599 Ext 87438 > Fax: 64 9 373 7484 > > http://www.fmhs.auckland.ac.nz/sms/biru/ > |
|
Lloyd Donaldson |
|
|
In reply to this post by Jacqueline Ross
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Jacqui We have a 63x 1.2NA water, a 63x 1.4NA oil and a 63x 1.3NA glycerol objectives. I would rank them in that order in terms of image quality - the glycerol lens is the best although the oil is close behind. The water lens is not so good but that may relate to our samples which have a high refractive index and hence water is not ideal. If you want to try them out with your samples let me know - your samples are quite different from ours so could behave quite differently. Regards - Lloyd Dr Lloyd Donaldson Project Leader - Microscopy & Wood Identification Manufacturing and Biomaterials Scion - Forests, Products, Innovation 49 Sala Street, Rotorua 3010 Ph 07 343 5581 www.scionresearch.com -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Jacqui Ross Sent: Thursday, 28 June 2012 6:47 p.m. To: [hidden email] Subject: Silicone oil immersion lenses for live cell imaging ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear All, I'm currently trialling 2 silicone oil immersion lenses from Olympus (30x/1.05NA and 60x/1.3NA) for live cell imaging on our Olympus FV1000 confocal, which is on an upright fixed stage microscope. So far, they both seem excellent and offer the higher level of resolution I'm looking for. I am particularly interested in how these lenses compare with high NA water-immersion objective lenses. At the moment, we have long working distance water-immersion lenses or standard oil immersion so I don't have a high NA water-immersion objective lens to do my own comparison. If anyone has personal experience with both types of lenses, I would be keen to hear their view. I'm particularly interested in how the 60x/1.2NA water-immersion lens compares to the silicone oil immersion 60x/1.3NA, since I would like to buy one or the other. Kind regards, Jacqui Jacqueline Ross Biomedical Imaging Microscopist Biomedical Imaging Research Unit School of Medical Sciences Faculty of Medical & Health Sciences The University of Auckland Private Bag 92019 Auckland, NEW ZEALAND Tel: 64 9 373 7599 Ext 87438 Fax: 64 9 373 7484 http://www.fmhs.auckland.ac.nz/sms/biru/ This e-mail and any attachments may contain information which is confidential or subject to copyright. If you receive this e-mail in error, please delete it. Scion does not accept responsibility for anything in this e-mail which is not provided in the course of Scion's usual business or for any computer virus, data corruption, interference or delay arising from this e-mail. |
|
Zac Arrac Atelaz |
|
|
In reply to this post by Stefan Sokoll
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Stefan: I am almost sure that Sebastian is talking about a 5mm thick crystal to built the dichroic on to, so all the dichroics will have this thickness. Best regards > Date: Thu, 28 Jun 2012 15:04:44 +0200 > From: [hidden email] > Subject: Re: calibration of confocal spinning disk setup (yokogawa) > To: [hidden email] > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Sebastian, > > I am simply calculating the x/y fwhm of each particle. I already used different objectives, still > giving the same results, thus it's probably not a matter of the objective. Also we use no > aftermagnification. > > Unfortunately, I don't understand what you mean with using 5mm substrates on the dicroics. Maybe you > can explain this in more detail? > > Thanks! > Stefan > > Sebastian Rhode wrote: > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Hi Stefan, > > > > from my experience so far, there can be a couple of things go wrong. When I > > found asymmetric PSF, the most common problem were caused by: > > > > - the objective itself - Did you check the PSF of those beads using the same > > objective on different setups? > > > > - Do you use a CMount with built-in optics (Aftermagnification) since use a > > CCD with rather large pixels? > > > > - Dichroic - especially the CSU-X1 use very thin (0.5mm) dichroics, which > > will not be very flat (for this reason we use 5mm substrates for our > > dicroics) - so this might induce a curvature in one direction and finally > > lead to astigmatism > > > > By the way, what are the values you calculated? > > > > By the way, no commercial interests here. > > > > Cheers, Sebi > > > > Dr. Sebastian Rhode > > Project Manager Research & Development > > > > TILL Photonics GmbH > > an FEI Company > > > > > > > > On Thu, 28 Jun 2012 12:08:02 +0200, Stefan Sokoll <[hidden email]> > > wrote: > > > >> ***** > >> To join, leave or search the confocal microscopy listserv, go to: > >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >> ***** > >> > >> Hello, > >> > >> I am doing single particle tracking using an upright confocal spinning disk > > setup. The setup > >> basically consists of an Olympus TIRF100x oil objective, an Olympus > > microscope and on top a Yokogawa > >> spinning disk CSU-X1 with an Andor EMCCD camera. > >> > >> Analyzing the 3D PSF of latex beats I discover that the PSF is not > > symmetrical in xy direction, > >> instead the x expansion is roughly 50nm larger than in y direction. > >> > >> I wonder what might be the main cause for this effect? Tilted cover slip, > > adjustment of the devices, > >> the CSU itself or ...? So far I just put the devices on top of each other > > but due to the upright > >> setup I cannot check the symmetry of the laser beam e.g. over long distance > > at a wall. > >> Any ideas on how to best start calibrating the setup are very welcome. > > Thanks in advance! > >> Best, > >> Stefan > >> > >> -- > >> --------- > >> Dipl. Ing.-Inf. Stefan Sokoll > >> Lab. Molecular Physiology > >> Dept. Neurochemistry & Molecular Biology > >> Leibniz Institute for Neurobiology (LIN) > >> Brenneckestrasse 6 > >> 39118 Magdeburg > >> Germany > >> > >> Mail: [hidden email] > >> Tel.: +49 - 391 - 626393171 > > -- > --------- > Dipl. Ing.-Inf. Stefan Sokoll > Lab. Molecular Physiology > Dept. Neurochemistry & Molecular Biology > Leibniz Institute for Neurobiology (LIN) > Brenneckestrasse 6 > 39118 Magdeburg > Germany > > Mail: [hidden email] > Tel.: +49 - 391 - 626393171 |
|
Jacqueline Ross |
|
|
In reply to this post by Lloyd Donaldson
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Lloyd, Andrew, George, Thanks very much for your replies, which are very helpful. I also got some other comments off-list which were also much appreciated. Thanks Lloyd for the offer to test your lenses. I'll keep it in mind but I really want the comparison against the silicone oil immersion lens. Interesting to hear that the glycerol lens is the best for your specimens. Thanks Andrew for your comments and comparison of the two lenses. We have a large variety of specimens imaged on our systems including zebrafish so it would be good to offer increased resolution as well as improved depth. I wasn't aware of the "PSF-grade" lenses so I'll be contacting my local representative to ask about that! With regard to George's information, the information from Olympus on the refractive index matching is also the reason that I am considering this lens. Otherwise, I would have just purchased the water lens. Finally, in answer to your question Pascal, it is a bit trickier using these immersion lenses for live cell work on an upright microscope. We have to use a chamber with coverglass, i.e. you culture the cells on coverslips and then use a small chamber. There are a number of these available in the market currently. Kind regards, Jacqui Jacqueline Ross Biomedical Imaging Microscopist Biomedical Imaging Research Unit School of Medical Sciences Faculty of Medical & Health Sciences The University of Auckland Private Bag 92019 Auckland, NEW ZEALAND Tel: 64 9 373 7599 Ext 87438 Fax: 64 9 373 7484 http://www.fmhs.auckland.ac.nz/sms/biru/ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Lloyd Donaldson Sent: Friday, 29 June 2012 8:07 a.m. To: [hidden email] Subject: Re: Silicone oil immersion lenses for live cell imaging ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Jacqui We have a 63x 1.2NA water, a 63x 1.4NA oil and a 63x 1.3NA glycerol objectives. I would rank them in that order in terms of image quality - the glycerol lens is the best although the oil is close behind. The water lens is not so good but that may relate to our samples which have a high refractive index and hence water is not ideal. If you want to try them out with your samples let me know - your samples are quite different from ours so could behave quite differently. Regards - Lloyd Dr Lloyd Donaldson Project Leader - Microscopy & Wood Identification Manufacturing and Biomaterials Scion - Forests, Products, Innovation 49 Sala Street, Rotorua 3010 Ph 07 343 5581 www.scionresearch.com -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Jacqui Ross Sent: Thursday, 28 June 2012 6:47 p.m. To: [hidden email] Subject: Silicone oil immersion lenses for live cell imaging ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear All, I'm currently trialling 2 silicone oil immersion lenses from Olympus (30x/1.05NA and 60x/1.3NA) for live cell imaging on our Olympus FV1000 confocal, which is on an upright fixed stage microscope. So far, they both seem excellent and offer the higher level of resolution I'm looking for. I am particularly interested in how these lenses compare with high NA water-immersion objective lenses. At the moment, we have long working distance water-immersion lenses or standard oil immersion so I don't have a high NA water-immersion objective lens to do my own comparison. If anyone has personal experience with both types of lenses, I would be keen to hear their view. I'm particularly interested in how the 60x/1.2NA water-immersion lens compares to the silicone oil immersion 60x/1.3NA, since I would like to buy one or the other. Kind regards, Jacqui Jacqueline Ross Biomedical Imaging Microscopist Biomedical Imaging Research Unit School of Medical Sciences Faculty of Medical & Health Sciences The University of Auckland Private Bag 92019 Auckland, NEW ZEALAND Tel: 64 9 373 7599 Ext 87438 Fax: 64 9 373 7484 http://www.fmhs.auckland.ac.nz/sms/biru/ This e-mail and any attachments may contain information which is confidential or subject to copyright. If you receive this e-mail in error, please delete it. Scion does not accept responsibility for anything in this e-mail which is not provided in the course of Scion's usual business or for any computer virus, data corruption, interference or delay arising from this e-mail. |
|
Robin Battye-2 |
|
|
In reply to this post by Stefan Sokoll
Sebastian is talking about the TILL spinning disc, not the Yokogawa spinning disc. The dichroics will be different between these two devices. With Yokogawa the dichroics is placed between the pinhole disc and the microlens disc and needs to meet specific dimensions, one of those includes a thin substrate on which to sputter you coatings. This thin substrate can be prone to bending, although steps can be taken to minimize this. The TILL system has a single disc and the dichroics is in free space so that you can make it as thick as you choose.
If you would like to discuss further details please contact me offline. Cheers Robin Robin Battye, M.Sc., Ph. D. Confocal Product Manager/ Technical Sales Specialist Mobile: (416) 302-5934 Spectral Applied Research 9078 Leslie St., Unit 11 Richmond Hill, Ontario L4B 3L8 Office: (905) 326-5040 ext. 406 Fax: (905) 326-5041 www.spectral.ca We believe that performance comes before Compromise. -------- Original message -------- Subject: Re: calibration of confocal spinning disk setup (yokogawa) From: Zac Arrac Atelaz <[hidden email]> To: [hidden email] CC: Re: calibration of confocal spinning disk setup (yokogawa) ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Stefan: I am almost sure that Sebastian is talking about a 5mm thick crystal to built the dichroic on to, so all the dichroics will have this thickness. Best regards > Date: Thu, 28 Jun 2012 15:04:44 +0200 > From: [hidden email] > Subject: Re: calibration of confocal spinning disk setup (yokogawa) > To: [hidden email] > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Sebastian, > > I am simply calculating the x/y fwhm of each particle. I already used different objectives, still > giving the same results, thus it's probably not a matter of the objective. Also we use no > aftermagnification. > > Unfortunately, I don't understand what you mean with using 5mm substrates on the dicroics. Maybe you > can explain this in more detail? > > Thanks! > Stefan > > Sebastian Rhode wrote: > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Hi Stefan, > > > > from my experience so far, there can be a couple of things go wrong. When I > > found asymmetric PSF, the most common problem were caused by: > > > > - the objective itself - Did you check the PSF of those beads using the same > > objective on different setups? > > > > - Do you use a CMount with built-in optics (Aftermagnification) since use a > > CCD with rather large pixels? > > > > - Dichroic - especially the CSU-X1 use very thin (0.5mm) dichroics, which > > will not be very flat (for this reason we use 5mm substrates for our > > dicroics) - so this might induce a curvature in one direction and finally > > lead to astigmatism > > > > By the way, what are the values you calculated? > > > > By the way, no commercial interests here. > > > > Cheers, Sebi > > > > Dr. Sebastian Rhode > > Project Manager Research & Development > > > > TILL Photonics GmbH > > an FEI Company > > > > > > > > On Thu, 28 Jun 2012 12:08:02 +0200, Stefan Sokoll <[hidden email]> > > wrote: > > > >> ***** > >> To join, leave or search the confocal microscopy listserv, go to: > >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >> ***** > >> > >> Hello, > >> > >> I am doing single particle tracking using an upright confocal spinning disk > > setup. The setup > >> basically consists of an Olympus TIRF100x oil objective, an Olympus > > microscope and on top a Yokogawa > >> spinning disk CSU-X1 with an Andor EMCCD camera. > >> > >> Analyzing the 3D PSF of latex beats I discover that the PSF is not > > symmetrical in xy direction, > >> instead the x expansion is roughly 50nm larger than in y direction. > >> > >> I wonder what might be the main cause for this effect? Tilted cover slip, > > adjustment of the devices, > >> the CSU itself or ...? So far I just put the devices on top of each other > > but due to the upright > >> setup I cannot check the symmetry of the laser beam e.g. over long distance > > at a wall. > >> Any ideas on how to best start calibrating the setup are very welcome. > > Thanks in advance! > >> Best, > >> Stefan > >> > >> -- > >> --------- > >> Dipl. Ing.-Inf. Stefan Sokoll > >> Lab. Molecular Physiology > >> Dept. Neurochemistry & Molecular Biology > >> Leibniz Institute for Neurobiology (LIN) > >> Brenneckestrasse 6 > >> 39118 Magdeburg > >> Germany > >> > >> Mail: [hidden email] > >> Tel.: +49 - 391 - 626393171 > > -- > --------- > Dipl. Ing.-Inf. Stefan Sokoll > Lab. Molecular Physiology > Dept. Neurochemistry & Molecular Biology > Leibniz Institute for Neurobiology (LIN) > Brenneckestrasse 6 > 39118 Magdeburg > Germany > > Mail: [hidden email] > Tel.: +49 - 391 - 626393171 |
|
|
In reply to this post by George McNamara
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello, This is a totally non-confocal and likely trivial question. I got a small USB microscope camera and while I see how it can be controlled by the OS under Win XP, I am at a total loss with Win 7 Pro. a) is there a build in webcam viewer like in XP b) is there a way to link such a camera to ImageJ of Fiji or maybe there is a freeware standalone program controlling such a camera. Thanks in advance, Michal -- * Dr. Michal Opas Professor Department of Laboratory Medicine and Pathobiology University of Toronto 1 King's College Circle Medical Sciences Building, room 6326 Toronto, Ontario, M5S 1A8 Canada **°°°°°°°°°°°°°** phone: (416) 978-8947 (laboratory) (416) 971-2140 (office) fax: (416) 978-5959 e*-*mail: **[hidden email] <mailto:[hidden email]>*** *WWW**:** **http://www.utoronto.ca/mocell * |
|
George McNamara |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Michal, You can setup "XP mode" inside Windows 7 ... or download and setup VMware Player and install XP in that. ImageJ ... Looking at the Micro-Manager.org list, there are TWAIN and "OpenCVgrabber" in the camera list http://valelab.ucsf.edu/~MM/MMwiki/index.php/Device_Support George p.s. suggestion for the micro-manager folks ... how about interfacing with the Canon Hack Development kit, recently mentioned on the listserv http://chdk.wikia.com/wiki/CHDK I see it CHDK was written up in IEEE Spectrum http://spectrum.ieee.org/geek-life/hands-on/using-the-canon-hack-development-kit/0/ You (mm folks( should consider writing an article for IEEE Spectrum. Also for Journal of Biomolecular Technologies. On 7/2/2012 8:48 PM, Michal Opas wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello, > > This is a totally non-confocal and likely trivial question. I got a > small USB microscope camera and while I see how it can be controlled > by the OS under Win XP, I am at a total loss with Win 7 Pro. > a) is there a build in webcam viewer like in XP > b) is there a way to link such a camera to ImageJ of Fiji or maybe > there is a freeware standalone program controlling such a camera. > > Thanks in advance, > > Michal > |
|
Michal Opas |
|
|
In reply to this post by George McNamara
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello Listers, Forgive my ignorance but did anyone used plastic coverslips for immunofluorescence? If so would standard immerion oil work? Thanks in advance! Michal * <http://www.utoronto.ca/mocell>* |
|
leoncio vergara |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** I havent use plastic coverslips for anything in a long time, but working with IBIDI chambers, it seems not immersion oils are compatible. The company provide a list of brands and products that are compatible with their slides. So Iguess the answer would be, not all immersion oils are the same, some may not be compatible. Leoncio ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of Michal Opas [[hidden email]] Sent: Wednesday, January 16, 2013 8:58 AM To: [hidden email] Subject: plastic coverslips & immersion imaging ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello Listers, Forgive my ignorance but did anyone used plastic coverslips for immunofluorescence? If so would standard immerion oil work? Thanks in advance! Michal * <http://www.utoronto.ca/mocell>* |
|
Richard E. Edelmann |
|
|
In reply to this post by Michal Opas
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Yes, for one series of projects we did use plastic coverslips (studying biofilm formation on plastic substrate). And yes normal cargill immersion oil worked. The biggest issue we had with the plastic coverslips was thickness - thickness was terrible. And often would exceed the 100um working distance (WD) of the lens we were using. (i.e. the 170um optical coverslip thickness + 100um WD) and thus the samples could not be focused. Took a bit to discover this, then we took to using a micrometer to measure the coverslips and moved to a 150um WD lens. On 16 Jan 2013 at 9:58, Michal Opas wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** > > Hello Listers, > > Forgive my ignorance but did anyone used plastic coverslips for > immunofluorescence? If so would standard immerion oil work? > Thanks in advance! > > Michal * > <http://www.utoronto.ca/mocell>* Richard E. Edelmann, Ph.D., Director Center for Advanced Microscopy & Imaging 9C Upham Hall Miami University, Oxford, OH 45056 Ph: 513.529.5712 Fax: 513.529.4243 E-mail: [hidden email] http://www.cami.muohio.edu |