Renato Mortara |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear friends, A colleague is trying to image mouse brain sections that were radioactively labeled (I guess 35S-Methionine) then, immersed in a photographic emulsion to develop particular sites. The sections were then counter stained with toluidine or similar (bluish) histological dye. The final result ('bright spots, easily seen under a 20x objective) can be visualized under dark field microscopy. I wonder if anyone has been able to image through some kind of reflection mode under confocal microscopy. Thanks ! Renato Mortara Dr. Renato Arruda Mortara Parasitology Division Escola Paulista de Medicina - UNIFESP Rua Botucatu, 862 6th floor 04023-062 São Paulo SP Brazil Phone: 55 11 55798306 www.ecb.epm.br/~ramortara |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Renato, I have (years ago) done silver-enhanced gold immunolabelling in reflection mode confocal microscopy. It stands out brilliantly. Labelling that is invisible in transmitted light is very strong in confocal. It gives you an extra channel over and above what you are imaging in fluorescence, with no cross-talk. I also discovered that the images are so 'different' that referees don't like them! Guy Guy Cox, Honorary Associate Professor School of Medical Sciences Australian Centre for Microscopy and Microanalysis, Madsen, F09, University of Sydney, NSW 2006 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Renato Mortara Sent: Monday, 25 November 2013 9:30 PM To: [hidden email] Subject: Silver enhanced autoradiography / dark field under confocal ? ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear friends, A colleague is trying to image mouse brain sections that were radioactively labeled (I guess 35S-Methionine) then, immersed in a photographic emulsion to develop particular sites. The sections were then counter stained with toluidine or similar (bluish) histological dye. The final result ('bright spots, easily seen under a 20x objective) can be visualized under dark field microscopy. I wonder if anyone has been able to image through some kind of reflection mode under confocal microscopy. Thanks ! Renato Mortara Dr. Renato Arruda Mortara Parasitology Division Escola Paulista de Medicina - UNIFESP Rua Botucatu, 862 6th floor 04023-062 São Paulo SP Brazil Phone: 55 11 55798306 www.ecb.epm.br/~ramortara |
|
George McNamara |
|
|
In reply to this post by Renato Mortara
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Renato, Yes, reflection confocal microscopy works great for this. You may need to disable a reflection suppression option to get maximum signal (hopefully a readily visible checkbox in the GUI - for example, the Leica LAS AF option is harder to find than Zeiss ZEN option). Evaluate each of your laser lines for best contrast and least noise. Evaluate different combinations of laser power and detector gain (ex. high laser power, low gain, assuming the laser line is not noisy). Probably simplest to leave the pinhole at 1.0 Airy units, though smaller might result in even better resolution. Hopefully they have controls - there are usually some background grains that need to be counted. Grains distributed in Z may or may not be due to 35S. Silver grain size may not be proportional to the amount of 35S in a given volume - and grain count may be linear over a limited range (though may not be practical to evaluate this). If all the laser lines perform well, you can choose the channel that gives the best counterstain image in the transmitted light detector. Note that some counterstains - notably eosin - are also fluorescent. If you / your colleague want a color transmitted light image, do this with 3 laser lines, such as 458 (Ar), 561 (DPSS), 633 (HeNe). The laser lines do not need to be perfectly "RGB", just close. To speed up imaging, you could just acquire two color channels and use one twice, such as 633 nm for R, and 488 nm for G and B. If the silver grains also show up in the transmitted light images, not a big deal: you have them in high contrast in the reflection confocal image. There was a recent article (Biotechniques?) whose whole point was that they were able to use LacZ beta galactosidase (BCIP or NBT/BCIP) product fluorescence confocal images to identify what black voids were product and what were irrelevant junk from the specimen. For image acquisition, more frame averaging is better - since silver grains will not photobleach, you can average as much as you want (if this is a core facility, you could maximize revenue by using maximum averaging). Hopefully your confocal microscope software include deconvolution (if you have Leica LAS AF try the different settings). If not, the Bruce and Butte NVidia CUDA based GPU deconvolution software http://www.opticsinfobase.org/oe/fulltext.cfm?uri=oe-21-4-4766&id=249375 <http://www.opticsinfobase.org/oe/fulltext.cfm?uri=oe-21-4-4766&id=249375> is fast and no charge for academic license - just need to complete and send to Manish the end user license agreement http://tcell.stanford.edu/software.html If you use it, do not just accept the -1 (automatic) default for number of iterations - evaluate different settings. I typically use 100 iterations on a Quadro 2000 card (about 6 seconds for 512x512x32 planes). I just ordered the new GeForce GTX 780 Ti (3 Gb card, about $700), which I expect will enable me to deconvolve full FLASH4.0 (2048x2048 pixels) fields of view. In addition to X and Y being a power of two (beneficial for FFTs), Manish and I recently exchanged emails where they reminded me that Z should also be a power of 2 for both speed and GPU ram efficiency (probably true of all deconvolution software). Sincerely, George On 11/25/2013 4:29 AM, Renato Mortara wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear friends, > > > > A colleague is trying to image mouse brain sections that were radioactively > labeled (I guess 35S-Methionine) then, immersed in a photographic emulsion > to develop particular sites. > > > > The sections were then counter stained with toluidine or similar (bluish) > histological dye. The final result ('bright spots, easily seen under a 20x > objective) can be visualized under dark field microscopy. > > > > I wonder if anyone has been able to image through some kind of reflection > mode under confocal microscopy. > > > > Thanks ! > > > > Renato Mortara > > > > Dr. Renato Arruda Mortara > > Parasitology Division > > Escola Paulista de Medicina - UNIFESP > > Rua Botucatu, 862 6th floor > > 04023-062 > > São Paulo SP Brazil > > Phone: 55 11 55798306 > > www.ecb.epm.br/~ramortara > > -- George McNamara, Ph.D. Single Cells Analyst L.J.N. Cooper Lab University of Texas M.D. Anderson Cancer Center Houston, TX 77054 Tattletales http://works.bepress.com/gmcnamara/26/ |
|
|
In reply to this post by Renato Mortara
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Back in the early 1990s we used reflectance confocal to image in situ. Also, enhanced immunogold and lead ATPase. No one has asked for it recently, but even after a decade it must still work. =========================================================================== Michael Cammer, Microscopy Core & Dustin Lab , Skirball Institute, NYU Langone Medical Center Cell: 914-309-3270 Lab: 212-263-3208 http://ocs.med.nyu.edu/microscopy & http://www.med.nyu.edu/skirball-lab/dustinlab/ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Renato Mortara Sent: Monday, November 25, 2013 5:30 AM To: [hidden email] Subject: Silver enhanced autoradiography / dark field under confocal ? ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear friends, A colleague is trying to image mouse brain sections that were radioactively labeled (I guess 35S-Methionine) then, immersed in a photographic emulsion to develop particular sites. The sections were then counter stained with toluidine or similar (bluish) histological dye. The final result ('bright spots, easily seen under a 20x objective) can be visualized under dark field microscopy. I wonder if anyone has been able to image through some kind of reflection mode under confocal microscopy. Thanks ! Renato Mortara Dr. Renato Arruda Mortara Parasitology Division Escola Paulista de Medicina - UNIFESP Rua Botucatu, 862 6th floor 04023-062 São Paulo SP Brazil Phone: 55 11 55798306 www.ecb.epm.br/~ramortara ------------------------------------------------------------ This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email. ================================= |
|
Jacqueline Ross |
|
|
In reply to this post by Renato Mortara
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Renato, I also imaged the same type of specimens - bovine cartilage explants, silver grains in photographic emulsion using reflectance confocal on a Leica TCS 4D confocal microscope. We published a couple of papers using this technique combined with immunostaining for type VI collagen together with biochemical analysis of the various tissue fractions. It worked brilliantly for assessing localisation. I can send PDFs of the papers offline if your colleague is interested. Kind regards, Jacqui Jacqueline Ross Biomedical Imaging Microscopist Biomedical Imaging Research Unit School of Medical Sciences Faculty of Medical & Health Sciences The University of Auckland Private Bag 92019 Auckland 1142, NEW ZEALAND Tel: 64 9 923 7438 Fax: 64 9 373 7484 http://www.fmhs.auckland.ac.nz/sms/biru/ -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Renato Mortara Sent: Monday, 25 November 2013 11:30 p.m. To: [hidden email] Subject: Silver enhanced autoradiography / dark field under confocal ? ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear friends, A colleague is trying to image mouse brain sections that were radioactively labeled (I guess 35S-Methionine) then, immersed in a photographic emulsion to develop particular sites. The sections were then counter stained with toluidine or similar (bluish) histological dye. The final result ('bright spots, easily seen under a 20x objective) can be visualized under dark field microscopy. I wonder if anyone has been able to image through some kind of reflection mode under confocal microscopy. Thanks ! Renato Mortara Dr. Renato Arruda Mortara Parasitology Division Escola Paulista de Medicina - UNIFESP Rua Botucatu, 862 6th floor 04023-062 São Paulo SP Brazil Phone: 55 11 55798306 www.ecb.epm.br/~ramortara |
|
George McNamara |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Paddock and Eliceiri, chapter 2 in Paddock 2013, page 33, figure 15 illustrates the benefit of reflection confocal microscopy of silver grains compared to transmitted light darkfield or transmitted light brightfield. http://www.amazon.com/Confocal-Microscopy-Methods-Protocols-Molecular/dp/1588293513 Figure was previously published in Paddock 2002 Biotechnoqies, figure 3, http://www.biotechniques.com/multimedia/archive/00014/paddock322_14678a.pdf On 11/26/2013 10:54 PM, Jacqui Ross wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hi Renato, > > I also imaged the same type of specimens - bovine cartilage explants, silver grains in photographic emulsion using reflectance confocal on a Leica TCS 4D confocal microscope. We published a couple of papers using this technique combined with immunostaining for type VI collagen together with biochemical analysis of the various tissue fractions. > > It worked brilliantly for assessing localisation. > > I can send PDFs of the papers offline if your colleague is interested. > > Kind regards, > > Jacqui > > > Jacqueline Ross > Biomedical Imaging Microscopist > Biomedical Imaging Research Unit > School of Medical Sciences > Faculty of Medical& Health Sciences > The University of Auckland > Private Bag 92019 > Auckland 1142, NEW ZEALAND > > Tel: 64 9 923 7438 > Fax: 64 9 373 7484 > > http://www.fmhs.auckland.ac.nz/sms/biru/ > > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Renato Mortara > Sent: Monday, 25 November 2013 11:30 p.m. > To: [hidden email] > Subject: Silver enhanced autoradiography / dark field under confocal ? > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear friends, > > > > A colleague is trying to image mouse brain sections that were radioactively labeled (I guess 35S-Methionine) then, immersed in a photographic emulsion to develop particular sites. > > > > The sections were then counter stained with toluidine or similar (bluish) histological dye. The final result ('bright spots, easily seen under a 20x > objective) can be visualized under dark field microscopy. > > > > I wonder if anyone has been able to image through some kind of reflection mode under confocal microscopy. > > > > Thanks ! > > > > Renato Mortara > > > > Dr. Renato Arruda Mortara > > Parasitology Division > > Escola Paulista de Medicina - UNIFESP > > Rua Botucatu, 862 6th floor > > 04023-062 > > São Paulo SP Brazil > > Phone: 55 11 55798306 > > www.ecb.epm.br/~ramortara > > -- George McNamara, Ph.D. Single Cells Analyst L.J.N. Cooper Lab University of Texas M.D. Anderson Cancer Center Houston, TX 77054 Tattletales http://works.bepress.com/gmcnamara/26/ |