|
Rosemary.White |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, (sorry for cross-posting) I know this has been discussed before, just wondered if there were any new thoughts. We're moving twice, first into a refurbished basement, second time into a new building. The basement has a sprinkler system in case of fire, unlike our current building which just has alarms (it's a single storey brick building). I imagine the new building will have some sort of sprinkler system too. What precautions, if any, do people take to protect against the unlikely showering of your confocal or EM? Are there any relatively simple alternatives? (One alternative is to have fire-doors and walls for each room but that is prohibitive.) thanks much, Rosemary Dr Rosemary White CSIRO Plant Industry (for 4 more days only) GPO Box 1600 Canberra, ACT 2601 T 02 6246 5475 F 02 6246 5334 E [hidden email] |
|
Haller, Edward |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi, Rosemary, I fought (and lost) with the Fire Marshall and building Engineers when they built my lab. I designed the lab from ground up, and was never asked about sprinklers. There are no other buildings on our campus with sprinklers, and I was not informed that sprinklers were going to be installed in my labs when they were being built, nor was I asked where I wanted the heads installed. When I walked in the door to see the labs (I was actually forbidden for a month to walk through the labs during construction) I discovered a sprinkler head directly above the spot where my TEM column would sit! After much wrangling (and griping and belly-aching on the part of engineers and construction crew), I was at least able to insist that the head be moved to a side wall in the room, away from the microscope column, so that if the head ever leaks, it will not drip on the column, and so, if it is ever bumped, we would have a prayer of a chance to cover the thing and protect the microscope. I'd like to build a box to cover the head, but we get inspections from time to time, or, more ideally, hire an outside plumber to cut the line. I have sprinklers in the rest of the lab, and fire extinguishers right outside the door. I don't need a sprinkler head in my TEM room! I also have a head in my SEM room, but it's further away from my scope. I could probably throw a plastic dropcloth on that scope and save it. (As a side note, this same construction crew installed two chilled water spigots UNDER my wall-mounted desk, in the knee space, in my office, and installed the lock to my TEM room on the INSIDE of the door. They weren't the brightest lightbulbs in the boxes!) Ed Edward Haller, Lab Manager University of South Florida Department of Integrative Biology Electron Microscopy Core SCA 110 4202 East Fowler Avenue Tampa, FL 33620 (813)974-2676 [hidden email] Office: ISA 1046 http://biology.usf.edu/ib/research/facilities/ ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of [hidden email] [[hidden email]] Sent: Thursday, June 26, 2014 6:05 AM To: [hidden email] Subject: Sprinklers in instrument rooms ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, (sorry for cross-posting) I know this has been discussed before, just wondered if there were any new thoughts. We're moving twice, first into a refurbished basement, second time into a new building. The basement has a sprinkler system in case of fire, unlike our current building which just has alarms (it's a single storey brick building). I imagine the new building will have some sort of sprinkler system too. What precautions, if any, do people take to protect against the unlikely showering of your confocal or EM? Are there any relatively simple alternatives? (One alternative is to have fire-doors and walls for each room but that is prohibitive.) thanks much, Rosemary Dr Rosemary White CSIRO Plant Industry (for 4 more days only) GPO Box 1600 Canberra, ACT 2601 T 02 6246 5475 F 02 6246 5334 E [hidden email] |
|
Feinstein, Timothy |
|
|
In reply to this post by Rosemary.White
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Rosemary, I knew a guy who made a corrugated plastic roof (slanted, with drainage) to protect his scope from water leakage due to floods and burst pipes. Sprinklers on the other hand strike me as a risky sword of damocles hanging over your stuff. If you need something to meet building codes, we use a compressed halon system. You trade danger to the scopes for suffocation risk, but it works on a sixty second timer so people have plenty of time to leave the room. We forbid earphones to make sure people can hear the alarm. Cheers, TF Timothy Feinstein, Ph.D. | Confocal Manager 333 Bostwick Ave., N.E., Grand Rapids, Michigan 49503 Phone: 616-234-5819 | Email: [hidden email] On 6/26/14, 6:05 AM, "[hidden email]" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://scanmail.trustwave.com/?c=129&d=0PCr08vv_PhM_NYM3Sbni5i1-7gOhUV1a9e >uOzpPog&u=http%3a%2f%2flists%2eumn%2eedu%2fcgi-bin%2fwa%3fA0%3dconfocalmic >roscopy >Post images on >http://scanmail.trustwave.com/?c=129&d=0PCr08vv_PhM_NYM3Sbni5i1-7gOhUV1a4K >vaTAb8g&u=http%3a%2f%2fwww%2eimgur%2ecom and include the link in your >posting. >***** > >Hi all, (sorry for cross-posting) > >I know this has been discussed before, just wondered if there were any >new thoughts. We're moving twice, first into a refurbished basement, >second time into a new building. The basement has a sprinkler system in >case of fire, unlike our current building which just has alarms (it's a >single storey brick building). I imagine the new building will have some >sort of sprinkler system too. > >What precautions, if any, do people take to protect against the unlikely >showering of your confocal or EM? >Are there any relatively simple alternatives? (One alternative is to have >fire-doors and walls for each room but that is prohibitive.) > >thanks much, >Rosemary > >Dr Rosemary White >CSIRO Plant Industry (for 4 more days only) >GPO Box 1600 >Canberra, ACT 2601 > >T 02 6246 5475 >F 02 6246 5334 >E [hidden email] |
|
|
In reply to this post by Haller, Edward
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** If fire code requires sprinklers, you get sprinklers. The best you can do is build a cage around the microscope as in the physiology labs. But don't think you can get away with hanging a tarp over the equipment. =========================================================================== Michael Cammer, Microscopy Core & Dustin Lab , Skirball Institute, NYU Langone Medical Center Cell: 914-309-3270 Lab: 212-263-3208 http://ocs.med.nyu.edu/microscopy & http://www.med.nyu.edu/skirball-lab/dustinlab/ ------------------------------------------------------------ This email message, including any attachments, is for the sole use of the intended recipient(s) and may contain information that is proprietary, confidential, and exempt from disclosure under applicable law. Any unauthorized review, use, disclosure, or distribution is prohibited. If you have received this email in error please notify the sender by return email and delete the original message. Please note, the recipient should check this email and any attachments for the presence of viruses. The organization accepts no liability for any damage caused by any virus transmitted by this email. ================================= |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** We do have sprinklers in our lab (building code). I use large overhead shelves suspended from the ceiling over all my rigs to hold lower priority equipment like computers. The shelves are two layered so a second higher shelf which is mostly empty covers the computers. If it 'rains' in the lab the shelves should deflect the bulk of the water away from the most critical components. The other thing we have is good insurance. Craig Brideau On Thu, Jun 26, 2014 at 11:16 AM, Cammer, Michael < [hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > If fire code requires sprinklers, you get sprinklers. The best you can do > is build a cage around the microscope as in the physiology labs. But don't > think you can get away with hanging a tarp over the equipment. > > =========================================================================== > Michael Cammer, Microscopy Core & Dustin Lab , Skirball Institute, NYU > Langone Medical Center > Cell: 914-309-3270 Lab: 212-263-3208 > http://ocs.med.nyu.edu/microscopy & > http://www.med.nyu.edu/skirball-lab/dustinlab/ > > > > ------------------------------------------------------------ > This email message, including any attachments, is for the sole use of the > intended recipient(s) and may contain information that is proprietary, > confidential, and exempt from disclosure under applicable law. Any > unauthorized review, use, disclosure, or distribution is prohibited. If you > have received this email in error please notify the sender by return email > and delete the original message. Please note, the recipient should check > this email and any attachments for the presence of viruses. The > organization accepts no liability for any damage caused by any virus > transmitted by this email. > ================================= > |
|
Martin Wessendorf-2 |
|
|
In reply to this post by Rosemary.White
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** As I understand it, most sprinkler systems go off only if there is fire (or a sufficiently high temperature--e.g., hot enough to boil alcohol) in the room--i.e., if a sprinkler goes off, you WANT it to go off--at that point, you have a fire in the room the water damage will be incidental. (The exception to this generalization seems to be "deluge" systems-- see: http://en.wikipedia.org/wiki/Sprinkler_Systems.) However, it might be possible to accidentally hit a sprinkler and set it off, if (for instance) someone were careless when moving equipment. If the ceilings (and sprinklers) are low and your equipment racks are tall, you might want to have wire guards placed around the sprinklers to protect them. Dr. Haller makes a good point about leakage--anyone know how common that is? My own sense is that the real danger is of a sprinkler going off on the floor *above *you. That could flood your space and cause needless damage. Tarps and shelves might shield you from that but might also prevent the sprinklers from properly drenching the room if there were a fire in your own equipment. Martin Wessendorf On 6/26/2014 5:05 AM, [hidden email] wrote: > Hi all, (sorry for cross-posting) > > I know this has been discussed before, just wondered if there were any new thoughts. We're moving twice, first into a refurbished basement, second time into a new building. The basement has a sprinkler system in case of fire, unlike our current building which just has alarms (it's a single storey brick building). I imagine the new building will have some sort of sprinkler system too. > > What precautions, if any, do people take to protect against the unlikely showering of your confocal or EM? > Are there any relatively simple alternatives? (One alternative is to have fire-doors and walls for each room but that is prohibitive.) > > thanks much, > Rosemary > > Dr Rosemary White > CSIRO Plant Industry (for 4 more days only) > GPO Box 1600 > Canberra, ACT 2601 -- Martin Wessendorf, Ph.D. office: (612) 626-0145 Assoc Prof, Dept Neuroscience lab: (612) 624-2991 University of Minnesota Preferred FAX: (612) 624-8118 6-145 Jackson Hall, 321 Church St. SE Dept Fax: (612) 626-5009 Minneapolis, MN 55455 e-mail: [hidden email] |
|
Oshel, Philip Eugene |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** There are also sprinkler systems using water that have empty pipes. The pipes to the sprinkler heads only fill with water when the system is activated. So no leaks. (This is supposed to be the system that goes in our new building. If the building is built right ...) Phil On 06/26/2014 15:06 , Martin Wessendorf wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > As I understand it, most sprinkler systems go off only if there is fire > (or a sufficiently high temperature--e.g., hot enough to boil alcohol) > in the room--i.e., if a sprinkler goes off, you WANT it to go off--at > that point, you have a fire in the room the water damage will be > incidental. (The exception to this generalization seems to be "deluge" > systems-- see: http://en.wikipedia.org/wiki/Sprinkler_Systems.) However, > it might be possible to accidentally hit a sprinkler and set it off, if > (for instance) someone were careless when moving equipment. If the > ceilings (and sprinklers) are low and your equipment racks are tall, you > might want to have wire guards placed around the sprinklers to protect > them. > > Dr. Haller makes a good point about leakage--anyone know how common that > is? My own sense is that the real danger is of a sprinkler going off on > the floor *above *you. That could flood your space and cause needless > damage. Tarps and shelves might shield you from that but might also > prevent the sprinklers from properly drenching the room if there were a > fire in your own equipment. > > Martin Wessendorf > > On 6/26/2014 5:05 AM, [hidden email] wrote: >> Hi all, (sorry for cross-posting) >> >> I know this has been discussed before, just wondered if there were any >> new thoughts. We're moving twice, first into a refurbished basement, >> second time into a new building. The basement has a sprinkler system >> in case of fire, unlike our current building which just has alarms >> (it's a single storey brick building). I imagine the new building will >> have some sort of sprinkler system too. >> >> What precautions, if any, do people take to protect against the >> unlikely showering of your confocal or EM? >> Are there any relatively simple alternatives? (One alternative is to >> have fire-doors and walls for each room but that is prohibitive.) >> >> thanks much, >> Rosemary >> >> Dr Rosemary White >> CSIRO Plant Industry (for 4 more days only) >> GPO Box 1600 >> Canberra, ACT 2601 > -- Philip Oshel Microscopy Facility Supervisor Biology Department 024C Brooks Hall Central Michigan University Mt. Pleasant, MI 48859 (989) 774-3576 |
|
Lemasters, John J. |
|
|
In reply to this post by Haller, Edward
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Because sprinkler heads are mandatory, we have put beach umbrellas over our microscopes. I have a colleague who lost several high end systems when a water line broke on a floor above his facility some years ago. The sprinkler heads are not the only thing to worry about. ----- John J. Lemasters, MD, PhD Professor and GlaxoSmithKline Distinguished Endowed Chair Director, Center for Cell Death, Injury & Regeneration Departments of Drug Discovery & Biomedical Sciences and Biochemistry & Molecular Biology Medical University of South Carolina DD504 Drug Discovery Building 70 President Street, MSC 140 Charleston, SC 29425 Email: [hidden email] -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Haller, Edward Sent: Thursday, June 26, 2014 9:40 AM To: [hidden email] Subject: Re: Sprinklers in instrument rooms ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi, Rosemary, I fought (and lost) with the Fire Marshall and building Engineers when they built my lab. I designed the lab from ground up, and was never asked about sprinklers. There are no other buildings on our campus with sprinklers, and I was not informed that sprinklers were going to be installed in my labs when they were being built, nor was I asked where I wanted the heads installed. When I walked in the door to see the labs (I was actually forbidden for a month to walk through the labs during construction) I discovered a sprinkler head directly above the spot where my TEM column would sit! After much wrangling (and griping and belly-aching on the part of engineers and construction crew), I was at least able to insist that the head be moved to a side wall in the room, away from the microscope column, so that if the head ever leaks, it will not drip on the column, and so, if it is ever bumped, we would have a prayer of a chance to cover the thing and protect the microscope. I'd like to build a box to cover the head, but we get inspections from time to time, or, more ideally, hire an outside plumber to cut the line. I have sprinklers in the rest of the lab, and fire extinguishers right outside the door. I don't need a sprinkler head in my TEM room! I also have a head in my SEM room, but it's further away from my scope. I could probably throw a plastic dropcloth on that scope and save it. (As a side note, this same construction crew installed two chilled water spigots UNDER my wall-mounted desk, in the knee space, in my office, and installed the lock to my TEM room on the INSIDE of the door. They weren't the brightest lightbulbs in the boxes!) Ed Edward Haller, Lab Manager University of South Florida Department of Integrative Biology Electron Microscopy Core SCA 110 4202 East Fowler Avenue Tampa, FL 33620 (813)974-2676 [hidden email] Office: ISA 1046 http://biology.usf.edu/ib/research/facilities/ ________________________________________ From: Confocal Microscopy List [[hidden email]] on behalf of [hidden email] [[hidden email]] Sent: Thursday, June 26, 2014 6:05 AM To: [hidden email] Subject: Sprinklers in instrument rooms ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi all, (sorry for cross-posting) I know this has been discussed before, just wondered if there were any new thoughts. We're moving twice, first into a refurbished basement, second time into a new building. The basement has a sprinkler system in case of fire, unlike our current building which just has alarms (it's a single storey brick building). I imagine the new building will have some sort of sprinkler system too. What precautions, if any, do people take to protect against the unlikely showering of your confocal or EM? Are there any relatively simple alternatives? (One alternative is to have fire-doors and walls for each room but that is prohibitive.) thanks much, Rosemary Dr Rosemary White CSIRO Plant Industry (for 4 more days only) GPO Box 1600 Canberra, ACT 2601 T 02 6246 5475 F 02 6246 5334 E [hidden email] |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I'm with John on this; it's not so much the sprinklers in your imaging room so much as the odds of a flood coming in from the floors above. Craig On Fri, Jun 27, 2014 at 8:23 AM, Lemasters, John J. <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Because sprinkler heads are mandatory, we have put beach umbrellas over > our microscopes. I have a colleague who lost several high end systems when > a water line broke on a floor above his facility some years ago. The > sprinkler heads are not the only thing to worry about. > > ----- > John J. Lemasters, MD, PhD > Professor and GlaxoSmithKline Distinguished Endowed Chair > Director, Center for Cell Death, Injury & Regeneration > Departments of Drug Discovery & Biomedical Sciences and Biochemistry & > Molecular Biology > Medical University of South Carolina > DD504 Drug Discovery Building > 70 President Street, MSC 140 > Charleston, SC 29425 > Email: [hidden email] > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of Haller, Edward > Sent: Thursday, June 26, 2014 9:40 AM > To: [hidden email] > Subject: Re: Sprinklers in instrument rooms > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi, Rosemary, > > I fought (and lost) with the Fire Marshall and building Engineers > when they built my lab. I designed the lab from ground up, and was never > asked about sprinklers. There are no other buildings on our campus with > sprinklers, and I was not informed that sprinklers were going to be > installed in my labs when they were being built, nor was I asked where I > wanted the heads installed. When I walked in the door to see the labs (I > was actually forbidden for a month to walk through the labs during > construction) I discovered a sprinkler head directly above the spot where > my TEM column would sit! After much wrangling (and griping and belly-aching > on the part of engineers and construction crew), I was at least able to > insist that the head be moved to a side wall in the room, away from the > microscope column, so that if the head ever leaks, it will not drip on the > column, and so, if it is ever bumped, we would have a prayer of a chance to > cover the thing and protect the microscope. I'd like to build a box to > cover the head, but we get inspections from time to time, or, more ideally, > hire an outside plumber to cut the line. I have sprinklers in the rest of > the lab, and fire extinguishers right outside the door. I don't need a > sprinkler head in my TEM room! I also have a head in my SEM room, but it's > further away from my scope. I could probably throw a plastic dropcloth on > that scope and save it. (As a side note, this same construction crew > installed two chilled water spigots UNDER my wall-mounted desk, in the knee > space, in my office, and installed the lock to my TEM room on the INSIDE of > the door. They weren't the brightest lightbulbs in the boxes!) > > Ed > > Edward Haller, Lab Manager > University of South Florida > Department of Integrative Biology > Electron Microscopy Core > SCA 110 > 4202 East Fowler Avenue > Tampa, FL 33620 > (813)974-2676 > [hidden email] > Office: ISA 1046 > http://biology.usf.edu/ib/research/facilities/ > > ________________________________________ > From: Confocal Microscopy List [[hidden email]] on > behalf of [hidden email] [[hidden email]] > Sent: Thursday, June 26, 2014 6:05 AM > To: [hidden email] > Subject: Sprinklers in instrument rooms > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi all, (sorry for cross-posting) > > I know this has been discussed before, just wondered if there were any new > thoughts. We're moving twice, first into a refurbished basement, second > time into a new building. The basement has a sprinkler system in case of > fire, unlike our current building which just has alarms (it's a single > storey brick building). I imagine the new building will have some sort of > sprinkler system too. > > What precautions, if any, do people take to protect against the unlikely > showering of your confocal or EM? > Are there any relatively simple alternatives? (One alternative is to have > fire-doors and walls for each room but that is prohibitive.) > > thanks much, > Rosemary > > Dr Rosemary White > CSIRO Plant Industry (for 4 more days only) GPO Box 1600 Canberra, ACT 2601 > > T 02 6246 5475 > F 02 6246 5334 > E [hidden email] > |
|
Douglas Richardson |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Our system requires both a smoke alarm and heat sensor to trip before water is delivered to the pipes. Day-to-day, there is no water in the pipes over our microscopes. -Doug On Fri, Jun 27, 2014 at 12:43 PM, Craig Brideau <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I'm with John on this; it's not so much the sprinklers in your imaging room > so much as the odds of a flood coming in from the floors above. > > Craig > > > On Fri, Jun 27, 2014 at 8:23 AM, Lemasters, John J. <[hidden email]> > wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Because sprinkler heads are mandatory, we have put beach umbrellas over > > our microscopes. I have a colleague who lost several high end systems > when > > a water line broke on a floor above his facility some years ago. The > > sprinkler heads are not the only thing to worry about. > > > > ----- > > John J. Lemasters, MD, PhD > > Professor and GlaxoSmithKline Distinguished Endowed Chair > > Director, Center for Cell Death, Injury & Regeneration > > Departments of Drug Discovery & Biomedical Sciences and Biochemistry & > > Molecular Biology > > Medical University of South Carolina > > DD504 Drug Discovery Building > > 70 President Street, MSC 140 > > Charleston, SC 29425 > > Email: [hidden email] > > > > -----Original Message----- > > From: Confocal Microscopy List [mailto:[hidden email]] > > On Behalf Of Haller, Edward > > Sent: Thursday, June 26, 2014 9:40 AM > > To: [hidden email] > > Subject: Re: Sprinklers in instrument rooms > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Hi, Rosemary, > > > > I fought (and lost) with the Fire Marshall and building Engineers > > when they built my lab. I designed the lab from ground up, and was never > > asked about sprinklers. There are no other buildings on our campus with > > sprinklers, and I was not informed that sprinklers were going to be > > installed in my labs when they were being built, nor was I asked where I > > wanted the heads installed. When I walked in the door to see the labs (I > > was actually forbidden for a month to walk through the labs during > > construction) I discovered a sprinkler head directly above the spot where > > my TEM column would sit! After much wrangling (and griping and > belly-aching > > on the part of engineers and construction crew), I was at least able to > > insist that the head be moved to a side wall in the room, away from the > > microscope column, so that if the head ever leaks, it will not drip on > the > > column, and so, if it is ever bumped, we would have a prayer of a chance > to > > cover the thing and protect the microscope. I'd like to build a box to > > cover the head, but we get inspections from time to time, or, more > ideally, > > hire an outside plumber to cut the line. I have sprinklers in the rest of > > the lab, and fire extinguishers right outside the door. I don't need a > > sprinkler head in my TEM room! I also have a head in my SEM room, but > it's > > further away from my scope. I could probably throw a plastic dropcloth on > > that scope and save it. (As a side note, this same construction crew > > installed two chilled water spigots UNDER my wall-mounted desk, in the > knee > > space, in my office, and installed the lock to my TEM room on the INSIDE > of > > the door. They weren't the brightest lightbulbs in the boxes!) > > > > Ed > > > > Edward Haller, Lab Manager > > University of South Florida > > Department of Integrative Biology > > Electron Microscopy Core > > SCA 110 > > 4202 East Fowler Avenue > > Tampa, FL 33620 > > (813)974-2676 > > [hidden email] > > Office: ISA 1046 > > http://biology.usf.edu/ib/research/facilities/ > > > > ________________________________________ > > From: Confocal Microscopy List [[hidden email]] on > > behalf of [hidden email] [[hidden email]] > > Sent: Thursday, June 26, 2014 6:05 AM > > To: [hidden email] > > Subject: Sprinklers in instrument rooms > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > Hi all, (sorry for cross-posting) > > > > I know this has been discussed before, just wondered if there were any > new > > thoughts. We're moving twice, first into a refurbished basement, second > > time into a new building. The basement has a sprinkler system in case of > > fire, unlike our current building which just has alarms (it's a single > > storey brick building). I imagine the new building will have some sort of > > sprinkler system too. > > > > What precautions, if any, do people take to protect against the unlikely > > showering of your confocal or EM? > > Are there any relatively simple alternatives? (One alternative is to have > > fire-doors and walls for each room but that is prohibitive.) > > > > thanks much, > > Rosemary > > > > Dr Rosemary White > > CSIRO Plant Industry (for 4 more days only) GPO Box 1600 Canberra, ACT > 2601 > > > > T 02 6246 5475 > > F 02 6246 5334 > > E [hidden email] > > > |
|
Feinstein, Timothy |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Keeping the sprinkler pipes empty is good, but I have seen lab floods come in an impressive number of ways. These include a nearby water main break (twice), pipes freezing five floors above, old building water pipes giving way and a staff member leaving a double distilled water tap running to fill a carboy and going home for the weekend. If you want to be triple safe and you have enough headroom, some sort of backup water protection is never a bad idea. That goes triple in a very old building. Also try to keep expensive things at least a couple of inches off the floor. All the best, TF Timothy Feinstein, Ph.D. | Confocal Manager 333 Bostwick Ave., N.E., Grand Rapids, Michigan 49503 Phone: 616-234-5819 | Email: [hidden email] Timothy Feinstein, Ph.D. | Confocal Manager 333 Bostwick Ave., N.E., Grand Rapids, Michigan 49503 Phone: 616-234-5819 | Email: [hidden email] On 6/27/14, 1:37 PM, "Douglas Richardson" <[hidden email]> wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://scanmail.trustwave.com/?c=129&d=_qut04nva1srzT9d7xTv86IByczaldOX_Vq >T8yBEhg&u=http%3a%2f%2flists%2eumn%2eedu%2fcgi-bin%2fwa%3fA0%3dconfocalmic >roscopy >Post images on >http://scanmail.trustwave.com/?c=129&d=_qut04nva1srzT9d7xTv86IByczaldOX_Q- >SoSoQ1g&u=http%3a%2f%2fwww%2eimgur%2ecom and include the link in your >posting. >***** > >Our system requires both a smoke alarm and heat sensor to trip before >water >is delivered to the pipes. Day-to-day, there is no water in the pipes >over >our microscopes. > >-Doug > > >On Fri, Jun 27, 2014 at 12:43 PM, Craig Brideau <[hidden email]> >wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> >>http://scanmail.trustwave.com/?c=129&d=_qut04nva1srzT9d7xTv86IByczaldOX_V >>qT8yBEhg&u=http%3a%2f%2flists%2eumn%2eedu%2fcgi-bin%2fwa%3fA0%3dconfocalm >>icroscopy >> Post images on >>http://scanmail.trustwave.com/?c=129&d=_qut04nva1srzT9d7xTv86IByczaldOX_Q >>-SoSoQ1g&u=http%3a%2f%2fwww%2eimgur%2ecom and include the link in your >>posting. >> ***** >> >> I'm with John on this; it's not so much the sprinklers in your imaging >>room >> so much as the odds of a flood coming in from the floors above. >> >> Craig >> >> >> On Fri, Jun 27, 2014 at 8:23 AM, Lemasters, John J. <[hidden email]> >> wrote: >> >> > ***** >> > To join, leave or search the confocal microscopy listserv, go to: >> > >>http://scanmail.trustwave.com/?c=129&d=_qut04nva1srzT9d7xTv86IByczaldOX_V >>qT8yBEhg&u=http%3a%2f%2flists%2eumn%2eedu%2fcgi-bin%2fwa%3fA0%3dconfocalm >>icroscopy >> > Post images on >>http://scanmail.trustwave.com/?c=129&d=_qut04nva1srzT9d7xTv86IByczaldOX_Q >>-SoSoQ1g&u=http%3a%2f%2fwww%2eimgur%2ecom and include the link in your >> posting. >> > ***** >> > >> > Because sprinkler heads are mandatory, we have put beach umbrellas >>over >> > our microscopes. I have a colleague who lost several high end systems >> when >> > a water line broke on a floor above his facility some years ago. The >> > sprinkler heads are not the only thing to worry about. >> > >> > ----- >> > John J. Lemasters, MD, PhD >> > Professor and GlaxoSmithKline Distinguished Endowed Chair >> > Director, Center for Cell Death, Injury & Regeneration >> > Departments of Drug Discovery & Biomedical Sciences and Biochemistry & >> > Molecular Biology >> > Medical University of South Carolina >> > DD504 Drug Discovery Building >> > 70 President Street, MSC 140 >> > Charleston, SC 29425 >> > Email: [hidden email] >> > >> > -----Original Message----- >> > From: Confocal Microscopy List >>[mailto:[hidden email]] >> > On Behalf Of Haller, Edward >> > Sent: Thursday, June 26, 2014 9:40 AM >> > To: [hidden email] >> > Subject: Re: Sprinklers in instrument rooms >> > >> > ***** >> > To join, leave or search the confocal microscopy listserv, go to: >> > >>http://scanmail.trustwave.com/?c=129&d=_qut04nva1srzT9d7xTv86IByczaldOX_V >>qT8yBEhg&u=http%3a%2f%2flists%2eumn%2eedu%2fcgi-bin%2fwa%3fA0%3dconfocalm >>icroscopy >> > Post images on >>http://scanmail.trustwave.com/?c=129&d=_qut04nva1srzT9d7xTv86IByczaldOX_Q >>-SoSoQ1g&u=http%3a%2f%2fwww%2eimgur%2ecom and include the link in your >> posting. >> > ***** >> > >> > Hi, Rosemary, >> > >> > I fought (and lost) with the Fire Marshall and building Engineers >> > when they built my lab. I designed the lab from ground up, and was >>never >> > asked about sprinklers. There are no other buildings on our campus >>with >> > sprinklers, and I was not informed that sprinklers were going to be >> > installed in my labs when they were being built, nor was I asked >>where I >> > wanted the heads installed. When I walked in the door to see the labs >>(I >> > was actually forbidden for a month to walk through the labs during >> > construction) I discovered a sprinkler head directly above the spot >>where >> > my TEM column would sit! After much wrangling (and griping and >> belly-aching >> > on the part of engineers and construction crew), I was at least able >>to >> > insist that the head be moved to a side wall in the room, away from >>the >> > microscope column, so that if the head ever leaks, it will not drip on >> the >> > column, and so, if it is ever bumped, we would have a prayer of a >>chance >> to >> > cover the thing and protect the microscope. I'd like to build a box to >> > cover the head, but we get inspections from time to time, or, more >> ideally, >> > hire an outside plumber to cut the line. I have sprinklers in the >>rest of >> > the lab, and fire extinguishers right outside the door. I don't need a >> > sprinkler head in my TEM room! I also have a head in my SEM room, but >> it's >> > further away from my scope. I could probably throw a plastic >>dropcloth on >> > that scope and save it. (As a side note, this same construction crew >> > installed two chilled water spigots UNDER my wall-mounted desk, in the >> knee >> > space, in my office, and installed the lock to my TEM room on the >>INSIDE >> of >> > the door. They weren't the brightest lightbulbs in the boxes!) >> > >> > Ed >> > >> > Edward Haller, Lab Manager >> > University of South Florida >> > Department of Integrative Biology >> > Electron Microscopy Core >> > SCA 110 >> > 4202 East Fowler Avenue >> > Tampa, FL 33620 >> > (813)974-2676 >> > [hidden email] >> > Office: ISA 1046 >> > >>http://scanmail.trustwave.com/?c=129&d=_qut04nva1srzT9d7xTv86IByczaldOX_V >>2doy8S3g&u=http%3a%2f%2fbiology%2eusf%2eedu%2fib%2fresearch%2ffacilities% >>2f >> > >> > ________________________________________ >> > From: Confocal Microscopy List [[hidden email]] on >> > behalf of [hidden email] [[hidden email]] >> > Sent: Thursday, June 26, 2014 6:05 AM >> > To: [hidden email] >> > Subject: Sprinklers in instrument rooms >> > >> > ***** >> > To join, leave or search the confocal microscopy listserv, go to: >> > >>http://scanmail.trustwave.com/?c=129&d=_qut04nva1srzT9d7xTv86IByczaldOX_V >>qT8yBEhg&u=http%3a%2f%2flists%2eumn%2eedu%2fcgi-bin%2fwa%3fA0%3dconfocalm >>icroscopy >> > Post images on >>http://scanmail.trustwave.com/?c=129&d=_qut04nva1srzT9d7xTv86IByczaldOX_Q >>-SoSoQ1g&u=http%3a%2f%2fwww%2eimgur%2ecom and include the link in your >> posting. >> > ***** >> > >> > Hi all, (sorry for cross-posting) >> > >> > I know this has been discussed before, just wondered if there were any >> new >> > thoughts. We're moving twice, first into a refurbished basement, >>second >> > time into a new building. The basement has a sprinkler system in case >>of >> > fire, unlike our current building which just has alarms (it's a single >> > storey brick building). I imagine the new building will have some >>sort of >> > sprinkler system too. >> > >> > What precautions, if any, do people take to protect against the >>unlikely >> > showering of your confocal or EM? >> > Are there any relatively simple alternatives? (One alternative is to >>have >> > fire-doors and walls for each room but that is prohibitive.) >> > >> > thanks much, >> > Rosemary >> > >> > Dr Rosemary White >> > CSIRO Plant Industry (for 4 more days only) GPO Box 1600 Canberra, ACT >> 2601 >> > >> > T 02 6246 5475 >> > F 02 6246 5334 >> > E [hidden email] >> > >> |
|
Barbara Foster |
|
|
In reply to this post by Craig Brideau
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** We've always made it a practice to have ... and USE the plastic covers on our light microscopes. There have been too many times when such a simple step would have avoided disaster. Good hunting! Barbara Foster, President & Chief Consultant Microscopy/Microscopy Education* www.MicroscopyEducation.com *A subsidiary of The Microscopy & Imaging Place, Inc. 7101 Royal Glen Trail, Suite A McKinney, TX 75070 P: 972-924-5310 F: 214-592-0277 MME is currently scheduling courses for now and through Fall 2014. Call us today for a free training evaluation. At 12:53 PM 6/27/2014, you wrote: >***** >To join, leave or search the confocal microscopy listserv, go to: >http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >Post images on http://www.imgur.com and include the link in your posting. >***** > >I'm with John on this; it's not so much the sprinklers in your imaging room >so much as the odds of a flood coming in from the floors above. > >Craig > > >On Fri, Jun 27, 2014 at 8:23 AM, Lemasters, John J. <[hidden email]> >wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your posting. > > ***** > > > > Because sprinkler heads are mandatory, we have put beach umbrellas over > > our microscopes. I have a colleague who lost several high end systems when > > a water line broke on a floor above his facility some years ago. The > > sprinkler heads are not the only thing to worry about. > > > > ----- > > John J. Lemasters, MD, PhD > > Professor and GlaxoSmithKline Distinguished Endowed Chair > > Director, Center for Cell Death, Injury & Regeneration > > Departments of Drug Discovery & Biomedical Sciences and Biochemistry & > > Molecular Biology > > Medical University of South Carolina > > DD504 Drug Discovery Building > > 70 President Street, MSC 140 > > Charleston, SC 29425 > > Email: [hidden email] > > > > -----Original Message----- > > From: Confocal Microscopy List [mailto:[hidden email]] > > On Behalf Of Haller, Edward > > Sent: Thursday, June 26, 2014 9:40 AM > > To: [hidden email] > > Subject: Re: Sprinklers in instrument rooms > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your posting. > > ***** > > > > Hi, Rosemary, > > > > I fought (and lost) with the Fire Marshall and building Engineers > > when they built my lab. I designed the lab from ground up, and was never > > asked about sprinklers. There are no other buildings on our campus with > > sprinklers, and I was not informed that sprinklers were going to be > > installed in my labs when they were being built, nor was I asked where I > > wanted the heads installed. When I walked in the door to see the labs (I > > was actually forbidden for a month to walk through the labs during > > construction) I discovered a sprinkler head directly above the spot where > > my TEM column would sit! After much wrangling (and griping and belly-aching > > on the part of engineers and construction crew), I was at least able to > > insist that the head be moved to a side wall in the room, away from the > > microscope column, so that if the head ever leaks, it will not drip on the > > column, and so, if it is ever bumped, we would have a prayer of a chance to > > cover the thing and protect the microscope. I'd like to build a box to > > cover the head, but we get inspections from time to time, or, more ideally, > > hire an outside plumber to cut the line. I have sprinklers in the rest of > > the lab, and fire extinguishers right outside the door. I don't need a > > sprinkler head in my TEM room! I also have a head in my SEM room, but it's > > further away from my scope. I could probably throw a plastic dropcloth on > > that scope and save it. (As a side note, this same construction crew > > installed two chilled water spigots UNDER my wall-mounted desk, in the knee > > space, in my office, and installed the lock to my TEM room on the INSIDE of > > the door. They weren't the brightest lightbulbs in the boxes!) > > > > Ed > > > > Edward Haller, Lab Manager > > University of South Florida > > Department of Integrative Biology > > Electron Microscopy Core > > SCA 110 > > 4202 East Fowler Avenue > > Tampa, FL 33620 > > (813)974-2676 > > [hidden email] > > Office: ISA 1046 > > http://biology.usf.edu/ib/research/facilities/ > > > > ________________________________________ > > From: Confocal Microscopy List [[hidden email]] on > > behalf of [hidden email] [[hidden email]] > > Sent: Thursday, June 26, 2014 6:05 AM > > To: [hidden email] > > Subject: Sprinklers in instrument rooms > > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your posting. > > ***** > > > > Hi all, (sorry for cross-posting) > > > > I know this has been discussed before, just wondered if there were any new > > thoughts. We're moving twice, first into a refurbished basement, second > > time into a new building. The basement has a sprinkler system in case of > > fire, unlike our current building which just has alarms (it's a single > > storey brick building). I imagine the new building will have some sort of > > sprinkler system too. > > > > What precautions, if any, do people take to protect against the unlikely > > showering of your confocal or EM? > > Are there any relatively simple alternatives? (One alternative is to have > > fire-doors and walls for each room but that is prohibitive.) > > > > thanks much, > > Rosemary > > > > Dr Rosemary White > > CSIRO Plant Industry (for 4 more days only) GPO Box 1600 Canberra, ACT 2601 > > > > T 02 6246 5475 > > F 02 6246 5334 > > E [hidden email] > > |
| Free forum by Nabble | Edit this page |