To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear confocalists, we have a very strange microscopy problem, that is not directly related to confocal microscopy. Based upon the huge expertise on this list however, I hope that someone may have stumbled upon a similar problem or at least has some suggestions for the cause of the problem. Currently we are trying to do widefield imaging of endogenous fluorophores (NADH, FAD). For this, we use a 365/50nm excitation filter and a 480/30nm emission bandpass. When we try to do multiposition imaging (automated stage from Prior Scientific controlled by an old Optiscan controller) we observe a strange sinus-like modulation in the signal (ONLY with the mentioned UV excitation filterset) that has an amplitude of approximately 10-15% of the initial signal. Cycle time for the sinusoidal artifact is approximately 10-15min. This artifact ONLY appears, when the stage is actually in use (single-position imaging yields a stable signal). The whole setup is driven by the most recent Micromanager release which is (accept for the observed perturbation) working just fine. So far, I can exclude an optical problem (imprecise repositioning of the stage and similar) - placing one of those fluorescent chroma-slides directly on the objective and repeating the experiments yields the very same results (sinusoidal artifact that vanishes when the stage is not in use). So far I have tried to attach grounding cables to virtually every piece of equipment including the mercury lamp and the camera (without any success in getting rid of this artifact). Relevant equipment is a Nikon Ti-E with PFS (does not make a difference, whether it is in the lightpath or not), an Hamamatsu Orca ER-G CCD camera (exposure times are all in the range from 200-400ms) and a standard Nikon HBO with an Osram HBO103w/2 (tried changing it against the equivalent lamp from a Zeiss Axiovert200m without any changes). Does anyone have any idea on how to tackle the problem? We are currently running out of ideas. King regards André -- Dr. André Bernardini Institute of Physiology University of Duisburg-Essen University Hospital of Essen Hufelandstr. 55 D-45147 Essen, Germany Ph ++49 201 723 4607 Fax ++49 201 723 4648 [hidden email] www.uni-due.de/physiologie |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Andre, Can it be a temperature artefact? Both NADH and FAD autofluorescence are temperature dependent. Regards, João -- João Lagarto, Ph.D Instituto Gulbenkian de Ciência, Oeiras, Portugal -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Andre Bernardini Sent: Monday, October 12, 2015 9:35 To: [hidden email] Subject: Strange microscopy problem ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear confocalists, we have a very strange microscopy problem, that is not directly related to confocal microscopy. Based upon the huge expertise on this list however, I hope that someone may have stumbled upon a similar problem or at least has some suggestions for the cause of the problem. Currently we are trying to do widefield imaging of endogenous fluorophores (NADH, FAD). For this, we use a 365/50nm excitation filter and a 480/30nm emission bandpass. When we try to do multiposition imaging (automated stage from Prior Scientific controlled by an old Optiscan controller) we observe a strange sinus-like modulation in the signal (ONLY with the mentioned UV excitation filterset) that has an amplitude of approximately 10-15% of the initial signal. Cycle time for the sinusoidal artifact is approximately 10-15min. This artifact ONLY appears, when the stage is actually in use (single-position imaging yields a stable signal). The whole setup is driven by the most recent Micromanager release which is (accept for the observed perturbation) working just fine. So far, I can exclude an optical problem (imprecise repositioning of the stage and similar) - placing one of those fluorescent chroma-slides directly on the objective and repeating the experiments yields the very same results (sinusoidal artifact that vanishes when the stage is not in use). So far I have tried to attach grounding cables to virtually every piece of equipment including the mercury lamp and the camera (without any success in getting rid of this artifact). Relevant equipment is a Nikon Ti-E with PFS (does not make a difference, whether it is in the lightpath or not), an Hamamatsu Orca ER-G CCD camera (exposure times are all in the range from 200-400ms) and a standard Nikon HBO with an Osram HBO103w/2 (tried changing it against the equivalent lamp from a Zeiss Axiovert200m without any changes). Does anyone have any idea on how to tackle the problem? We are currently running out of ideas. King regards André -- Dr. André Bernardini Institute of Physiology University of Duisburg-Essen University Hospital of Essen Hufelandstr. 55 D-45147 Essen, Germany Ph ++49 201 723 4607 Fax ++49 201 723 4648 [hidden email] www.uni-due.de/physiologie ----- No virus found in this message. Checked by AVG - www.avg.com Version: 2015.0.6172 / Virus Database: 4435/10804 - Release Date: 10/12/15 |
|
Andre Bernardini |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi João, thank you for your suggestion. I can exclude a temperature-artifact. The sample resides in a heated on-stage-incubator (37°C). Also a chroma-testslide, that lies directly on the objective, gives the very same artifact (doesn't matter whether on-stage-incubator heating is on or off), making a temperature-effect most very unlikely. Kind regards, André Dr. André Bernardini Institute of Physiology University of Duisburg-Essen University Hospital of Essen Hufelandstr. 55 D-45147 Essen, Germany Ph ++49 201 723 4607 Fax ++49 201 723 4648 [hidden email] www.uni-due.de/physiologie On 12.10.2015 11:17, Joao Lagarto wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Andre, > > Can it be a temperature artefact? Both NADH and FAD autofluorescence are temperature dependent. > > Regards, > João > > -- > João Lagarto, Ph.D > Instituto Gulbenkian de Ciência, > Oeiras, Portugal > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Andre Bernardini > Sent: Monday, October 12, 2015 9:35 > To: [hidden email] > Subject: Strange microscopy problem > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear confocalists, > > we have a very strange microscopy problem, that is not directly related to confocal microscopy. Based upon the huge expertise on this list however, I hope that someone may have stumbled upon a similar problem or at least has some suggestions for the cause of the problem. > > Currently we are trying to do widefield imaging of endogenous fluorophores (NADH, FAD). For this, we use a 365/50nm excitation filter and a 480/30nm emission bandpass. > > When we try to do multiposition imaging (automated stage from Prior Scientific controlled by an old Optiscan controller) we observe a strange sinus-like modulation in the signal (ONLY with the mentioned UV excitation filterset) that has an amplitude of approximately 10-15% of the initial signal. Cycle time for the sinusoidal artifact is approximately 10-15min. This artifact ONLY appears, when the stage is actually in use (single-position imaging yields a stable signal). The whole setup is driven by the most recent Micromanager release which is (accept for the observed perturbation) working just fine. > > So far, I can exclude an optical problem (imprecise repositioning of the stage and similar) - placing one of those fluorescent chroma-slides directly on the objective and repeating the experiments yields the very same results (sinusoidal artifact that vanishes when the stage is not in use). So far I have tried to attach grounding cables to virtually every piece of equipment including the mercury lamp and the camera (without any success in getting rid of this artifact). > > Relevant equipment is a Nikon Ti-E with PFS (does not make a difference, whether it is in the lightpath or not), an Hamamatsu Orca ER-G CCD camera (exposure times are all in the range from 200-400ms) and a standard Nikon HBO with an Osram HBO103w/2 (tried changing it against the equivalent lamp from a Zeiss Axiovert200m without any changes). > > Does anyone have any idea on how to tackle the problem? We are currently running out of ideas. > > King regards > André > > -- > Dr. André Bernardini > Institute of Physiology > University of Duisburg-Essen > University Hospital of Essen > Hufelandstr. 55 > D-45147 Essen, Germany > Ph ++49 201 723 4607 > Fax ++49 201 723 4648 > [hidden email] > www.uni-due.de/physiologie > > > > ----- > No virus found in this message. > Checked by AVG - www.avg.com > Version: 2015.0.6172 / Virus Database: 4435/10804 - Release Date: 10/12/15 > |
|
Steffen Dietzel |
|
|
In reply to this post by Andre Bernardini
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Andre, from what you are describing, probably either the excitation source or the detector or the z-position is fluctuating. I would try the following: - plug the mercury lamp into a different circuit, one that is not used by any other piece of this microscope, see if problem goes away. - Try the same for the camera, if it has a separate power supply. And maybe for the xy.motor - Do I read that correctly, that you don't see this fluctuation in green and red channels? not even a little in green? - exchange excitation source to something other than HBO. With a chroma-slide, you can get plenty of fluorescence also with a halogen lamp attached to the fluorescence light path, with longer excitation . - have you focused deep enough into the chroma-slide to exclude positioning problems in z? Do you keep the same focus over multiple exposures? - it still could be a temperature problem, but with the z-position or with temperature of the camera or the lamp. Although the latter two seem unlikely, as long as you don't feel temperature changes on your skin. Maybe the motor itself causes heat that induces these changes. Do you have one of those big climate chambers? If so, try the Chroma slide thing again, but with a fully opened climate box, to allow heat dissipation. Good hunting Steffen Am 12.10.2015 um 10:34 schrieb Andre Bernardini: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear confocalists, > > we have a very strange microscopy problem, that is not directly > related to confocal microscopy. Based upon the huge expertise on this > list however, I hope that someone may have stumbled upon a similar > problem or at least has some suggestions for the cause of the problem. > > Currently we are trying to do widefield imaging of endogenous > fluorophores (NADH, FAD). For this, we use a 365/50nm excitation > filter and a 480/30nm emission bandpass. > > When we try to do multiposition imaging (automated stage from Prior > Scientific controlled by an old Optiscan controller) we observe a > strange sinus-like modulation in the signal (ONLY with the mentioned > UV excitation filterset) that has an amplitude of approximately 10-15% > of the initial signal. Cycle time for the sinusoidal artifact is > approximately 10-15min. This artifact ONLY appears, when the stage is > actually in use (single-position imaging yields a stable signal). The > whole setup is driven by the most recent Micromanager release which is > (accept for the observed perturbation) working just fine. > > So far, I can exclude an optical problem (imprecise repositioning of > the stage and similar) - placing one of those fluorescent > chroma-slides directly on the objective and repeating the experiments > yields the very same results (sinusoidal artifact that vanishes when > the stage is not in use). So far I have tried to attach grounding > cables to virtually every piece of equipment including the mercury > lamp and the camera (without any success in getting rid of this > artifact). > > Relevant equipment is a Nikon Ti-E with PFS (does not make a > difference, whether it is in the lightpath or not), an Hamamatsu Orca > ER-G CCD camera (exposure times are all in the range from 200-400ms) > and a standard Nikon HBO with an Osram HBO103w/2 (tried changing it > against the equivalent lamp from a Zeiss Axiovert200m without any > changes). > > Does anyone have any idea on how to tackle the problem? We are > currently running out of ideas. > > King regards > André > > > -- > ------------------------------------------------------------ > Steffen Dietzel, PD Dr. rer. nat > Ludwig-Maximilians-Universität München > Biomedical Center (BMC) > Head of the Core Facility Bioimaging > > Großhaderner Straße 9 > D-82152 Planegg-Martinsried > Germany |
|
|
In reply to this post by Andre Bernardini
Dear Andre,
Currently I do not have any good ideas, but questions: 1) Is this 10-15 cycle dependent on the number of positions (locations) or the number of images (imaging speed)? 2) Do you see this modulation also if the stage movement is so small that you virtually stay at the same field of view? 3) Do you have a temperature control system? Greetings Gabor -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Andre Bernardini Sent: Montag, 12. Oktober 2015 10:35 To: [hidden email] Subject: Strange microscopy problem ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Dear confocalists, we have a very strange microscopy problem, that is not directly related to confocal microscopy. Based upon the huge expertise on this list however, I hope that someone may have stumbled upon a similar problem or at least has some suggestions for the cause of the problem. Currently we are trying to do widefield imaging of endogenous fluorophores (NADH, FAD). For this, we use a 365/50nm excitation filter and a 480/30nm emission bandpass. When we try to do multiposition imaging (automated stage from Prior Scientific controlled by an old Optiscan controller) we observe a strange sinus-like modulation in the signal (ONLY with the mentioned UV excitation filterset) that has an amplitude of approximately 10-15% of the initial signal. Cycle time for the sinusoidal artifact is approximately 10-15min. This artifact ONLY appears, when the stage is actually in use (single-position imaging yields a stable signal). The whole setup is driven by the most recent Micromanager release which is (accept for the observed perturbation) working just fine. So far, I can exclude an optical problem (imprecise repositioning of the stage and similar) - placing one of those fluorescent chroma-slides directly on the objective and repeating the experiments yields the very same results (sinusoidal artifact that vanishes when the stage is not in use). So far I have tried to attach grounding cables to virtually every piece of equipment including the mercury lamp and the camera (without any success in getting rid of this artifact). Relevant equipment is a Nikon Ti-E with PFS (does not make a difference, whether it is in the lightpath or not), an Hamamatsu Orca ER-G CCD camera (exposure times are all in the range from 200-400ms) and a standard Nikon HBO with an Osram HBO103w/2 (tried changing it against the equivalent lamp from a Zeiss Axiovert200m without any changes). Does anyone have any idea on how to tackle the problem? We are currently running out of ideas. King regards André -- Dr. André Bernardini Institute of Physiology University of Duisburg-Essen University Hospital of Essen Hufelandstr. 55 D-45147 Essen, Germany Ph ++49 201 723 4607 Fax ++49 201 723 4648 [hidden email] www.uni-due.de/physiologie |
|
|
In reply to this post by Steffen Dietzel
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Andre, Is the sinusoidal artifact in a single image every 10-15 minutes (sinusoidal variation of intensity across the single image) or is it appearing in the time-lapse as an increase in intensity of the whole image every 10-15 minutes? Eric Marino [hidden email] > On Oct 12, 2015, at 7:46 AM, Steffen Dietzel <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy <http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> > Post images on http://www.imgur.com <http://www.imgur.com/> and include the link in your posting. > ***** > > Andre, > > from what you are describing, probably either the excitation source or the detector or the z-position is fluctuating. > I would try the following: > - plug the mercury lamp into a different circuit, one that is not used by any other piece of this microscope, see if problem goes away. > - Try the same for the camera, if it has a separate power supply. And maybe for the xy.motor > - Do I read that correctly, that you don't see this fluctuation in green and red channels? not even a little in green? > - exchange excitation source to something other than HBO. With a chroma-slide, you can get plenty of fluorescence also with a halogen lamp attached to the fluorescence light path, with longer excitation . > - have you focused deep enough into the chroma-slide to exclude positioning problems in z? Do you keep the same focus over multiple exposures? > - it still could be a temperature problem, but with the z-position or with temperature of the camera or the lamp. Although the latter two seem unlikely, as long as you don't feel temperature changes on your skin. Maybe the motor itself causes heat that induces these changes. Do you have one of those big climate chambers? If so, try the Chroma slide thing again, but with a fully opened climate box, to allow heat dissipation. > > Good hunting > > Steffen > > > Am 12.10.2015 um 10:34 schrieb Andre Bernardini: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Dear confocalists, >> >> we have a very strange microscopy problem, that is not directly related to confocal microscopy. Based upon the huge expertise on this list however, I hope that someone may have stumbled upon a similar problem or at least has some suggestions for the cause of the problem. >> >> Currently we are trying to do widefield imaging of endogenous fluorophores (NADH, FAD). For this, we use a 365/50nm excitation filter and a 480/30nm emission bandpass. >> >> When we try to do multiposition imaging (automated stage from Prior Scientific controlled by an old Optiscan controller) we observe a strange sinus-like modulation in the signal (ONLY with the mentioned UV excitation filterset) that has an amplitude of approximately 10-15% of the initial signal. Cycle time for the sinusoidal artifact is approximately 10-15min. This artifact ONLY appears, when the stage is actually in use (single-position imaging yields a stable signal). The whole setup is driven by the most recent Micromanager release which is (accept for the observed perturbation) working just fine. >> >> So far, I can exclude an optical problem (imprecise repositioning of the stage and similar) - placing one of those fluorescent chroma-slides directly on the objective and repeating the experiments yields the very same results (sinusoidal artifact that vanishes when the stage is not in use). So far I have tried to attach grounding cables to virtually every piece of equipment including the mercury lamp and the camera (without any success in getting rid of this artifact). >> >> Relevant equipment is a Nikon Ti-E with PFS (does not make a difference, whether it is in the lightpath or not), an Hamamatsu Orca ER-G CCD camera (exposure times are all in the range from 200-400ms) and a standard Nikon HBO with an Osram HBO103w/2 (tried changing it against the equivalent lamp from a Zeiss Axiovert200m without any changes). >> >> Does anyone have any idea on how to tackle the problem? We are currently running out of ideas. >> >> King regards >> André >> >> >> -- >> ------------------------------------------------------------ >> Steffen Dietzel, PD Dr. rer. nat >> Ludwig-Maximilians-Universität München >> Biomedical Center (BMC) >> Head of the Core Facility Bioimaging >> >> Großhaderner Straße 9 >> D-82152 Planegg-Martinsried >> Germany |
|
|
In reply to this post by Andre Bernardini
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi André, The heated incubator does not rule out a temperature effect on its own. It will mostly depend on the temperature controller. If e.g. this operates on a periodic on-off switch, you can get fluctuations in the temperature that may affect the fluorescence signal. To rule out a temperature artefact, and given that both NADH and FAD fluorescence vary linearly with temperature, you could measure the temperature of the stage during the acquisition and try to correlate it with the fluorescence signal. Good luck, João -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Andre Bernardini Sent: Monday, October 12, 2015 10:28 To: [hidden email] Subject: Re: Strange microscopy problem ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi João, thank you for your suggestion. I can exclude a temperature-artifact. The sample resides in a heated on-stage-incubator (37°C). Also a chroma-testslide, that lies directly on the objective, gives the very same artifact (doesn't matter whether on-stage-incubator heating is on or off), making a temperature-effect most very unlikely. Kind regards, André Dr. André Bernardini Institute of Physiology University of Duisburg-Essen University Hospital of Essen Hufelandstr. 55 D-45147 Essen, Germany Ph ++49 201 723 4607 Fax ++49 201 723 4648 [hidden email] www.uni-due.de/physiologie On 12.10.2015 11:17, Joao Lagarto wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Andre, > > Can it be a temperature artefact? Both NADH and FAD autofluorescence are temperature dependent. > > Regards, > João > > -- > João Lagarto, Ph.D > Instituto Gulbenkian de Ciência, > Oeiras, Portugal > > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] On Behalf Of Andre > Bernardini > Sent: Monday, October 12, 2015 9:35 > To: [hidden email] > Subject: Strange microscopy problem > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear confocalists, > > we have a very strange microscopy problem, that is not directly related to confocal microscopy. Based upon the huge expertise on this list however, I hope that someone may have stumbled upon a similar problem or at least has some suggestions for the cause of the problem. > > Currently we are trying to do widefield imaging of endogenous fluorophores (NADH, FAD). For this, we use a 365/50nm excitation filter and a 480/30nm emission bandpass. > > When we try to do multiposition imaging (automated stage from Prior Scientific controlled by an old Optiscan controller) we observe a strange sinus-like modulation in the signal (ONLY with the mentioned UV excitation filterset) that has an amplitude of approximately 10-15% of the initial signal. Cycle time for the sinusoidal artifact is approximately 10-15min. This artifact ONLY appears, when the stage is actually in use (single-position imaging yields a stable signal). The whole setup is driven by the most recent Micromanager release which is (accept for the observed perturbation) working just fine. > > So far, I can exclude an optical problem (imprecise repositioning of the stage and similar) - placing one of those fluorescent chroma-slides directly on the objective and repeating the experiments yields the very same results (sinusoidal artifact that vanishes when the stage is not in use). So far I have tried to attach grounding cables to virtually every piece of equipment including the mercury lamp and the camera (without any success in getting rid of this artifact). > > Relevant equipment is a Nikon Ti-E with PFS (does not make a difference, whether it is in the lightpath or not), an Hamamatsu Orca ER-G CCD camera (exposure times are all in the range from 200-400ms) and a standard Nikon HBO with an Osram HBO103w/2 (tried changing it against the equivalent lamp from a Zeiss Axiovert200m without any changes). > > Does anyone have any idea on how to tackle the problem? We are currently running out of ideas. > > King regards > André > > -- > Dr. André Bernardini > Institute of Physiology > University of Duisburg-Essen > University Hospital of Essen > Hufelandstr. 55 > D-45147 Essen, Germany > Ph ++49 201 723 4607 > Fax ++49 201 723 4648 > [hidden email] > www.uni-due.de/physiologie > > > > ----- > No virus found in this message. > Checked by AVG - www.avg.com > Version: 2015.0.6172 / Virus Database: 4435/10804 - Release Date: > 10/12/15 > ----- No virus found in this message. Checked by AVG - www.avg.com Version: 2015.0.6172 / Virus Database: 4435/10804 - Release Date: 10/12/15 |
|
Andre Bernardini |
|
|
In reply to this post by Eric Marino
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Eric, it is the second one (increase / decrease over the whole image in a sinusoidal fashion). Imaging interval is one minute. I did not try setting the stage to really small xy-distances like Gabor suggested. However - it does not make a difference, whether the stage is actually mounted to the microscope or lying on the table beside it. I already tried Steffen's suggestion with the other power circuit (with all relevant devices) with no success. A z-Problem is also most unlikely because I observe the problem also with the chroma-testslide lying directly on the objective lens. In the meanwhile it turned out, that it is indeed some kind of electrical problem: I now provided proper grounding to the microscopes filterwheel (the one with the cubes for epifluorescence) and the problem seems virtually gone. I now observe a steady increase of fluorescence signal which could very well have other (physical) explanations (some kind of photoactivation / photoconversion in the chroma-testslide). I will have a closer look at this tomorrow. Does anyone now a reasonable explanation why a change in electric potential of the filterwheel could cause such an error? Anyway - thanks to all for your kind support! Kind regards André Dr. André Bernardini Institute of Physiology University of Duisburg-Essen University Hospital of Essen Hufelandstr. 55 D-45147 Essen, Germany Ph ++49 201 723 4607 Fax ++49 201 723 4648 [hidden email] www.uni-due.de/physiologie On 12.10.2015 16:21, Eric Marino wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Andre, > > Is the sinusoidal artifact in a single image every 10-15 minutes (sinusoidal variation of intensity across the single image) or is it appearing in the time-lapse as an increase in intensity of the whole image every 10-15 minutes? > > > Eric Marino > [hidden email] > > > >> On Oct 12, 2015, at 7:46 AM, Steffen Dietzel <[hidden email]> wrote: >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy <http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy> >> Post images on http://www.imgur.com <http://www.imgur.com/> and include the link in your posting. >> ***** >> >> Andre, >> >> from what you are describing, probably either the excitation source or the detector or the z-position is fluctuating. >> I would try the following: >> - plug the mercury lamp into a different circuit, one that is not used by any other piece of this microscope, see if problem goes away. >> - Try the same for the camera, if it has a separate power supply. And maybe for the xy.motor >> - Do I read that correctly, that you don't see this fluctuation in green and red channels? not even a little in green? >> - exchange excitation source to something other than HBO. With a chroma-slide, you can get plenty of fluorescence also with a halogen lamp attached to the fluorescence light path, with longer excitation . >> - have you focused deep enough into the chroma-slide to exclude positioning problems in z? Do you keep the same focus over multiple exposures? >> - it still could be a temperature problem, but with the z-position or with temperature of the camera or the lamp. Although the latter two seem unlikely, as long as you don't feel temperature changes on your skin. Maybe the motor itself causes heat that induces these changes. Do you have one of those big climate chambers? If so, try the Chroma slide thing again, but with a fully opened climate box, to allow heat dissipation. >> >> Good hunting >> >> Steffen >> >> >> Am 12.10.2015 um 10:34 schrieb Andre Bernardini: >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> Post images on http://www.imgur.com and include the link in your posting. >>> ***** >>> >>> Dear confocalists, >>> >>> we have a very strange microscopy problem, that is not directly related to confocal microscopy. Based upon the huge expertise on this list however, I hope that someone may have stumbled upon a similar problem or at least has some suggestions for the cause of the problem. >>> >>> Currently we are trying to do widefield imaging of endogenous fluorophores (NADH, FAD). For this, we use a 365/50nm excitation filter and a 480/30nm emission bandpass. >>> >>> When we try to do multiposition imaging (automated stage from Prior Scientific controlled by an old Optiscan controller) we observe a strange sinus-like modulation in the signal (ONLY with the mentioned UV excitation filterset) that has an amplitude of approximately 10-15% of the initial signal. Cycle time for the sinusoidal artifact is approximately 10-15min. This artifact ONLY appears, when the stage is actually in use (single-position imaging yields a stable signal). The whole setup is driven by the most recent Micromanager release which is (accept for the observed perturbation) working just fine. >>> >>> So far, I can exclude an optical problem (imprecise repositioning of the stage and similar) - placing one of those fluorescent chroma-slides directly on the objective and repeating the experiments yields the very same results (sinusoidal artifact that vanishes when the stage is not in use). So far I have tried to attach grounding cables to virtually every piece of equipment including the mercury lamp and the camera (without any success in getting rid of this artifact). >>> >>> Relevant equipment is a Nikon Ti-E with PFS (does not make a difference, whether it is in the lightpath or not), an Hamamatsu Orca ER-G CCD camera (exposure times are all in the range from 200-400ms) and a standard Nikon HBO with an Osram HBO103w/2 (tried changing it against the equivalent lamp from a Zeiss Axiovert200m without any changes). >>> >>> Does anyone have any idea on how to tackle the problem? We are currently running out of ideas. >>> >>> King regards >>> André >>> >>> >>> -- >>> ------------------------------------------------------------ >>> Steffen Dietzel, PD Dr. rer. nat >>> Ludwig-Maximilians-Universität München >>> Biomedical Center (BMC) >>> Head of the Core Facility Bioimaging >>> >>> Großhaderner Straße 9 >>> D-82152 Planegg-Martinsried >>> Germany |
|
|
In reply to this post by Joao Lagarto
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** We have seen a similar effect, on an incubated system with very similar timing and also not across all wavelengths. Our case was different as it was pinned down to an imperfect optic fibre for l;asers and the aircon heating/cooling the fibre itself. A change of fibre and shielding from room temperature effects fixed the problem. Other microscopes in the room have not exhibited the same problem. Therefore I would also try and discount any room aircon catching part of the light path and creating an intensity or optical shift. Best Pete Dr Peter O'Toole Head of Imaging and Cytometry Technology Facility Department of Biology (Area 15) University of York YORK YO10 5DD Tel : +44 (0)1904 328722 email : [hidden email] www.york.ac.uk/biology/tf Twitter: @YorkBioimaging EMAIL DISCLAIMER http://www.york.ac.uk/docs/disclaimer/email.htm -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Joao Lagarto Sent: 12 October 2015 15:29 To: [hidden email] Subject: Re: Strange microscopy problem ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi André, The heated incubator does not rule out a temperature effect on its own. It will mostly depend on the temperature controller. If e.g. this operates on a periodic on-off switch, you can get fluctuations in the temperature that may affect the fluorescence signal. To rule out a temperature artefact, and given that both NADH and FAD fluorescence vary linearly with temperature, you could measure the temperature of the stage during the acquisition and try to correlate it with the fluorescence signal. Good luck, João -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Andre Bernardini Sent: Monday, October 12, 2015 10:28 To: [hidden email] Subject: Re: Strange microscopy problem ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi João, thank you for your suggestion. I can exclude a temperature-artifact. The sample resides in a heated on-stage-incubator (37°C). Also a chroma-testslide, that lies directly on the objective, gives the very same artifact (doesn't matter whether on-stage-incubator heating is on or off), making a temperature-effect most very unlikely. Kind regards, André Dr. André Bernardini Institute of Physiology University of Duisburg-Essen University Hospital of Essen Hufelandstr. 55 D-45147 Essen, Germany Ph ++49 201 723 4607 Fax ++49 201 723 4648 [hidden email] www.uni-due.de/physiologie On 12.10.2015 11:17, Joao Lagarto wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Andre, > > Can it be a temperature artefact? Both NADH and FAD autofluorescence are temperature dependent. > > Regards, > João > > -- > João Lagarto, Ph.D > Instituto Gulbenkian de Ciência, > Oeiras, Portugal > > -----Original Message----- > From: Confocal Microscopy List > [mailto:[hidden email]] On Behalf Of Andre > Bernardini > Sent: Monday, October 12, 2015 9:35 > To: [hidden email] > Subject: Strange microscopy problem > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Dear confocalists, > > we have a very strange microscopy problem, that is not directly related to confocal microscopy. Based upon the huge expertise on this list however, I hope that someone may have stumbled upon a similar problem or at least has some suggestions for the cause of the problem. > > Currently we are trying to do widefield imaging of endogenous fluorophores (NADH, FAD). For this, we use a 365/50nm excitation filter and a 480/30nm emission bandpass. > > When we try to do multiposition imaging (automated stage from Prior Scientific controlled by an old Optiscan controller) we observe a strange sinus-like modulation in the signal (ONLY with the mentioned UV excitation filterset) that has an amplitude of approximately 10-15% of the initial signal. Cycle time for the sinusoidal artifact is approximately 10-15min. This artifact ONLY appears, when the stage is actually in use (single-position imaging yields a stable signal). The whole setup is driven by the most recent Micromanager release which is (accept for the observed perturbation) working just fine. > > So far, I can exclude an optical problem (imprecise repositioning of the stage and similar) - placing one of those fluorescent chroma-slides directly on the objective and repeating the experiments yields the very same results (sinusoidal artifact that vanishes when the stage is not in use). So far I have tried to attach grounding cables to virtually every piece of equipment including the mercury lamp and the camera (without any success in getting rid of this artifact). > > Relevant equipment is a Nikon Ti-E with PFS (does not make a difference, whether it is in the lightpath or not), an Hamamatsu Orca ER-G CCD camera (exposure times are all in the range from 200-400ms) and a standard Nikon HBO with an Osram HBO103w/2 (tried changing it against the equivalent lamp from a Zeiss Axiovert200m without any changes). > > Does anyone have any idea on how to tackle the problem? We are currently running out of ideas. > > King regards > André > > -- > Dr. André Bernardini > Institute of Physiology > University of Duisburg-Essen > University Hospital of Essen > Hufelandstr. 55 > D-45147 Essen, Germany > Ph ++49 201 723 4607 > Fax ++49 201 723 4648 > [hidden email] > www.uni-due.de/physiologie > > > > ----- > No virus found in this message. > Checked by AVG - www.avg.com > Version: 2015.0.6172 / Virus Database: 4435/10804 - Release Date: > 10/12/15 > ----- No virus found in this message. Checked by AVG - www.avg.com Version: 2015.0.6172 / Virus Database: 4435/10804 - Release Date: 10/12/15 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Yes Peter incubators have PID controllers and temperature is quite stable. HVAC systems vary as João was saying, with simple on/off control. Just try to switch the air conditioning off in the room and measure again. Nuno Moreno Scientific Facilities Instituto Gulbenkian de Ciencia > On 12 Oct 2015, at 15:39, Peter O'Toole <[hidden email]> wrote: > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > We have seen a similar effect, on an incubated system with very similar timing and also not across all wavelengths. Our case was different as it was pinned down to an imperfect optic fibre for l;asers and the aircon heating/cooling the fibre itself. A change of fibre and shielding from room temperature effects fixed the problem. Other microscopes in the room have not exhibited the same problem. Therefore I would also try and discount any room aircon catching part of the light path and creating an intensity or optical shift. > > Best > Pete > > Dr Peter O'Toole > Head of Imaging and Cytometry > Technology Facility > Department of Biology (Area 15) > University of York > YORK > YO10 5DD > > Tel : +44 (0)1904 328722 > email : [hidden email] > www.york.ac.uk/biology/tf > Twitter: @YorkBioimaging > > > > EMAIL DISCLAIMER http://www.york.ac.uk/docs/disclaimer/email.htm > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Joao Lagarto > Sent: 12 October 2015 15:29 > To: [hidden email] > Subject: Re: Strange microscopy problem > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi André, > > The heated incubator does not rule out a temperature effect on its own. It will mostly depend on the temperature controller. If e.g. this operates on a periodic on-off switch, you can get fluctuations in the temperature that may affect the fluorescence signal. To rule out a temperature artefact, and given that both NADH and FAD fluorescence vary linearly with temperature, you could measure the temperature of the stage during the acquisition and try to correlate it with the fluorescence signal. > > Good luck, > João > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Andre Bernardini > Sent: Monday, October 12, 2015 10:28 > To: [hidden email] > Subject: Re: Strange microscopy problem > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi João, > > thank you for your suggestion. > > I can exclude a temperature-artifact. The sample resides in a heated on-stage-incubator (37°C). > > Also a chroma-testslide, that lies directly on the objective, gives the very same artifact (doesn't matter whether on-stage-incubator heating is on or off), making a temperature-effect most very unlikely. > > Kind regards, > André > > Dr. André Bernardini > Institute of Physiology > University of Duisburg-Essen > University Hospital of Essen > Hufelandstr. 55 > D-45147 Essen, Germany > Ph ++49 201 723 4607 > Fax ++49 201 723 4648 > [hidden email] > www.uni-due.de/physiologie > > On 12.10.2015 11:17, Joao Lagarto wrote: >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Hi Andre, >> >> Can it be a temperature artefact? Both NADH and FAD autofluorescence are temperature dependent. >> >> Regards, >> João >> >> -- >> João Lagarto, Ph.D >> Instituto Gulbenkian de Ciência, >> Oeiras, Portugal >> >> -----Original Message----- >> From: Confocal Microscopy List >> [mailto:[hidden email]] On Behalf Of Andre >> Bernardini >> Sent: Monday, October 12, 2015 9:35 >> To: [hidden email] >> Subject: Strange microscopy problem >> >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> Dear confocalists, >> >> we have a very strange microscopy problem, that is not directly related to confocal microscopy. Based upon the huge expertise on this list however, I hope that someone may have stumbled upon a similar problem or at least has some suggestions for the cause of the problem. >> >> Currently we are trying to do widefield imaging of endogenous fluorophores (NADH, FAD). For this, we use a 365/50nm excitation filter and a 480/30nm emission bandpass. >> >> When we try to do multiposition imaging (automated stage from Prior Scientific controlled by an old Optiscan controller) we observe a strange sinus-like modulation in the signal (ONLY with the mentioned UV excitation filterset) that has an amplitude of approximately 10-15% of the initial signal. Cycle time for the sinusoidal artifact is approximately 10-15min. This artifact ONLY appears, when the stage is actually in use (single-position imaging yields a stable signal). The whole setup is driven by the most recent Micromanager release which is (accept for the observed perturbation) working just fine. >> >> So far, I can exclude an optical problem (imprecise repositioning of the stage and similar) - placing one of those fluorescent chroma-slides directly on the objective and repeating the experiments yields the very same results (sinusoidal artifact that vanishes when the stage is not in use). So far I have tried to attach grounding cables to virtually every piece of equipment including the mercury lamp and the camera (without any success in getting rid of this artifact). >> >> Relevant equipment is a Nikon Ti-E with PFS (does not make a difference, whether it is in the lightpath or not), an Hamamatsu Orca ER-G CCD camera (exposure times are all in the range from 200-400ms) and a standard Nikon HBO with an Osram HBO103w/2 (tried changing it against the equivalent lamp from a Zeiss Axiovert200m without any changes). >> >> Does anyone have any idea on how to tackle the problem? We are currently running out of ideas. >> >> King regards >> André >> >> -- >> Dr. André Bernardini >> Institute of Physiology >> University of Duisburg-Essen >> University Hospital of Essen >> Hufelandstr. 55 >> D-45147 Essen, Germany >> Ph ++49 201 723 4607 >> Fax ++49 201 723 4648 >> [hidden email] >> www.uni-due.de/physiologie >> >> >> >> ----- >> No virus found in this message. >> Checked by AVG - www.avg.com >> Version: 2015.0.6172 / Virus Database: 4435/10804 - Release Date: >> 10/12/15 >> > > > > ----- > No virus found in this message. > Checked by AVG - www.avg.com > Version: 2015.0.6172 / Virus Database: 4435/10804 - Release Date: 10/12/15 |
|
|
In reply to this post by Peter O'Toole
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** We are seeking maternity cover for one of our Senior Technicians in the Imaging & Cytometry Laboratory within the Bioscience Technology Facility, Department of Biology, University of York. The Imaging & Cytometry Laboratory is one of the six, specialist core laboratories within the Bioscience Technology Facility and focusses on all aspects of confocal and electron microscopy, flow cytometry, plus associated techniques. It has an enviable international reputation for training and instrument and method development. The work of the laboratory is wide and very varied and involves training, support and advice; experimental troubleshooting and equipment maintenance; method development. Our users are researchers from both the University of York and external organisations (companies, other universities, Government Institutes etc.). Working closely with the Laboratory Head and the rest of the team, you will be involved in delivering aspects of the work of the Laboratory to our users. This position represents an exciting opportunity to gain further experience in the application of microscopy and flow cytometry to a very wide range of challenging and interesting biological questions. You will be a team worker, have at least a degree in biology or related discipline, and a good knowledge and experience of flow cytometry or microscopy. The post is available at 60% Full Time Equivalent from 4 November 2015 and is anticipated to be available until 9 August 2016 in the first instance. Best Pete -- Dr Peter O'Toole Head of Imaging and Cytometry Technology Facility Department of Biology (Area 15) University of York YORK YO10 5DD Tel : +44 (0)1904 328722 Fax : +44 (0)1904 328804 email : [hidden email] www.york.ac.uk/biology/tf Times Higher Education University of the Year 2010 EMAIL DISCLAIMER http://www.york.ac.uk/docs/disclaimer/email.htm |