SUBSCRIBE CONFOCALMICROSCOPY Sara Smith |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear listers, We want to set up a calcium measurement system based on our old Nikon microscpe (Eclipse TE300). It has a very nice environmental chamber and equipped with Hamamatsu Digital Camera (C4742-95) and BioPoint filter wheel. We don't want something very fancy (like very high resolution or very high speed acquisition) and really want to make a full use of what we have now since we have a very limited budget. Based on my research, we might need to buy a filter cube suitable for calcium measurement (eg. Fura-2) and acquisition software with ratio imaging application. Anything else do we need? I was looking at Hamamatsu AQUACOSMOS/Basic System. It is a very nice system to do calcium measurement. However, it is quite pricy and sold as a whole package. I would really appreciate if anyone here with experience of setting up a calcium measurement system gives me some suggestion? Thank you, Yi |
|
Johannes Helm |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Good afternoon, an important decision - which most probably will be made by your budget - is whether you want to go for 1) a system that will be suitable for quantitative ratiometric measurements at minimum time delay between the measurements of the two channels, 2) a system that will be suitable for quantitative ratiometric measurements at which the time delay between the measurements of the two channels can be "long" - i.e. at least some seconds, 3) a system for qualitative measurements, only (at which there are some tricks to make even qualitative measurements a kind of semi-quantitative). I write this since you explicitely mentioned the dye substance "Fura-2" in your mail, which is an excitation ratiometric dye substance. Some comments: a) Before you consider ratiometric measurements at all, estimate your photon statistics. You will need quite a large number of photons to produce fully quantitative [Ca2+](x, y, z, t) images that are not drowned in noise, since the ratio imaging algorithm is strongly noise enhancing. Also, during your data analysis you should avoid using 8 bit or 12 bit values since truncation errors will have a terrible effect. Ref. for an analysis of that problem: Helm, Patwardhan, Manders, 1997, "A study of the precision of confocal, ratiometric, Fura-2-based [Ca2+]measurements", Cell Calcium 22(4):287-298 b) If you want to do ratiometric measurements, it will be cheaper to avoid excitation ratiometric dyes and go for an emission ratiometric dye, if possible, e.g. one of the Indoes. c) Whether excitation or emission ratiometric, you will - dependent on your measurement system - need either two filter cubes or one filter cube with two different filters for excitation or emission, which then will not be mounted in the cubes but somewhere else in the system (either on the illumination side for the excitation ratiometric system or on the emission side for the emission ratiometric system). d) Your budget might allow you to purchase one of these: http://www.photometrics.com/products/multichannel/dv2.php which will allow you to use your single camera as a two channel detector suitable even for fast- provided your rate of photons is large enough - emission ratiometric measurements. e) If ratiometry is beyond your financial horizon then you will most probably need to only purchase a suitable filtercube for your dye substance of choice. f) You may write your own data analysis software or check, e.g., ImageJ for Macros for your data analysis. Alternatively, you might inspect the market and find commercial data analysis software (possibly even combined with a suitable acquisition software). Since you have a Hamamatsu Camera, it might be a good idea to talk to your local Hamamatsu dealer about that. g) Definitely worth the investment even at a poor budget situation will be an electro mechanical shutter (e.g. the "Uniblitz" type by Vincent Associates, ref. www.uniblitz.com, or by JML, ref. http://www.jmloptical.com/pages/electronic-shutter-systems.aspx ) that can be triggered from your camera data acquisition software in order to minimize bleaching. Best wishes, Johannes -- P. Johannes Helm, M.Sc. PhD Seniorengineer CMBN University of Oslo Institute of Basic Medical Science Department of Anatomy Postboks 1105 - Blindern NO-0317 Oslo Voice: +47 228 51159 Fax: +47 228 51499 WWW: folk.uio.no/jhelm > ***** > To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear listers, > > We want to set up a calcium measurement system based on our old Nikon microscpe (Eclipse TE300). It has a very nice environmental chamber and equipped with Hamamatsu Digital Camera (C4742-95) and BioPoint filter wheel. > > We don't want something very fancy (like very high resolution or very high > speed acquisition) and really want to make a full use of what we have now > since we have a very limited budget. > Based on my research, we might need to buy a filter cube suitable for calcium > measurement (eg. Fura-2) and acquisition software with ratio imaging application. Anything else do we need? > > I was looking at Hamamatsu AQUACOSMOS/Basic System. It is a very nice system to do calcium measurement. However, it is quite pricy and sold as a > whole package. > > I would really appreciate if anyone here with experience of setting up a calcium measurement system gives me some suggestion? > > Thank you, > > Yi > |
|
Boswell, Carl A - (cboswell) |
|
|
In reply to this post by SUBSCRIBE CONFOCALMICROSCOPY Sara Smith
Hi Sara,
An additional item not mentioned in Johannes' excellent reply is the transmission limits of most lenses to UV wavelengths. If you are set on using Fura2, seriously consider a quartz or other high-UV transmission objective. Because so little 340nm light gets to the sample using "normal" lenses, the fluorescent signal received from the 340 excitation remains essentially unchanged with changes in intracellular calcium, rendering the sensitivity of your measurements very low. Good luck, C Carl A. Boswell 520-954-7053 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of SUBSCRIBE CONFOCALMICROSCOPY Sara Smith Sent: Friday, October 21, 2011 8:12 AM To: [hidden email] Subject: Suggestion for setting up Calcium measurement system ***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear listers, We want to set up a calcium measurement system based on our old Nikon microscpe (Eclipse TE300). It has a very nice environmental chamber and equipped with Hamamatsu Digital Camera (C4742-95) and BioPoint filter wheel. We don't want something very fancy (like very high resolution or very high speed acquisition) and really want to make a full use of what we have now since we have a very limited budget. Based on my research, we might need to buy a filter cube suitable for calcium measurement (eg. Fura-2) and acquisition software with ratio imaging application. Anything else do we need? I was looking at Hamamatsu AQUACOSMOS/Basic System. It is a very nice system to do calcium measurement. However, it is quite pricy and sold as a whole package. I would really appreciate if anyone here with experience of setting up a calcium measurement system gives me some suggestion? Thank you, Yi |
|
SUBSCRIBE CONFOCALMICROSCOPY Sara Smith |
|
|
In reply to this post by Johannes Helm
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Johannes, Thank you for all your inputs. I really appreciate your time and help. It is very helpful for us and all the information is valuable. For the measurement system, we want to have a least semi-quantitative feature. We are planning to measure the calcium concentration by using ratio metric dye. So the change of ratio of the intensity of two channels will present the change of calcium concentration. At this stage, we might not need to measure the exactly calcium concentration in uM. Although is semi-quantitative, we still want the time delay between the measurements of the two channels to be as short as possible. The reason I listed Fura-2 as an example is I used to use it and the microscope is equipped with a filter wheel, however we are open to any dye that can fulfill this mission. Our microscope is a wide-field fluorescent microscope not confocal. Is it still cheaper to go with an emission ratio metric dye? If this is the case, we can reconsider to use indoes instead of Fura-2. As for the system set up, I was so clueless and your email helps me a lot. It is very informative and guides me in the right direction. Thank you so much. I will read your paper and do more research about it. Thanks again! Yi On Fri, Oct 21, 2011 at 11:50 AM, P. Johannes Helm <[hidden email]>wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Good afternoon, > > an important decision - which most probably will be made by your budget - > is whether you want to go for > > 1) > a system that will be suitable for quantitative ratiometric measurements > at minimum time delay between the measurements of the two channels, > > 2) > a system that will be suitable for quantitative ratiometric measurements > at which the time delay between the measurements of the two channels can > be "long" - i.e. at least some seconds, > > 3) > a system for qualitative measurements, only (at which there are some > tricks to make even qualitative measurements a kind of semi-quantitative). > > I write this since you explicitely mentioned the dye substance "Fura-2" in > your mail, which is an excitation ratiometric dye substance. > > Some comments: > a) > Before you consider ratiometric measurements at all, estimate your photon > statistics. You will need quite a large number of photons to produce fully > quantitative [Ca2+](x, y, z, t) images that are not drowned in noise, > since the ratio imaging algorithm is strongly noise enhancing. Also, > during your data analysis you should avoid using 8 bit or 12 bit values > since truncation errors will have a terrible effect. > Ref. for an analysis of that problem: > Helm, Patwardhan, Manders, 1997, "A study of the precision of confocal, > ratiometric, Fura-2-based [Ca2+]measurements", Cell Calcium 22(4):287-298 > > > b) > If you want to do ratiometric measurements, it will be cheaper to avoid > excitation ratiometric dyes and go for an emission ratiometric dye, if > possible, e.g. one of the Indoes. > > > c) > Whether excitation or emission ratiometric, you will - dependent on your > measurement system - need either two filter cubes or one filter cube with > two different filters for excitation or emission, which then will not be > mounted in the cubes but somewhere else in the system (either on the > illumination side for the excitation ratiometric system or on the emission > side for the emission ratiometric system). > > d) > Your budget might allow you to purchase one of these: > http://www.photometrics.com/products/multichannel/dv2.php > which will allow you to use your single camera as a two channel detector > suitable even for fast- provided your rate of photons is large enough - > emission ratiometric measurements. > > e) > If ratiometry is beyond your financial horizon then you will most probably > need to only purchase a suitable filtercube for your dye substance of > choice. > > f) > You may write your own data analysis software or check, e.g., ImageJ for > Macros for your data analysis. Alternatively, you might inspect the market > and find commercial data analysis software (possibly even combined with a > suitable acquisition software). Since you have a Hamamatsu Camera, it > might be a good idea to talk to your local Hamamatsu dealer about that. > > g) > Definitely worth the investment even at a poor budget situation will be an > electro mechanical shutter (e.g. the "Uniblitz" type by Vincent > Associates, ref. www.uniblitz.com, or by JML, ref. > http://www.jmloptical.com/pages/electronic-shutter-systems.aspx ) that can > be triggered from your camera data acquisition software in order to > minimize bleaching. > > Best wishes, > > Johannes > > > -- > P. Johannes Helm, M.Sc. PhD > Seniorengineer > CMBN > University of Oslo > Institute of Basic Medical Science > Department of Anatomy > Postboks 1105 - Blindern > NO-0317 Oslo > > Voice: +47 228 51159 > Fax: +47 228 51499 > > WWW: folk.uio.no/jhelm > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > ***** > > > > Dear listers, > > > > We want to set up a calcium measurement system based on our old Nikon > microscpe (Eclipse TE300). It has a very nice environmental chamber and > equipped with Hamamatsu Digital Camera (C4742-95) and BioPoint filter > wheel. > > > > We don't want something very fancy (like very high resolution or very > high > > speed acquisition) and really want to make a full use of what we have > now > > since we have a very limited budget. > > Based on my research, we might need to buy a filter cube suitable for > calcium > > measurement (eg. Fura-2) and acquisition software with ratio imaging > application. Anything else do we need? > > > > I was looking at Hamamatsu AQUACOSMOS/Basic System. It is a very nice > system to do calcium measurement. However, it is quite pricy and sold as > a > > whole package. > > > > I would really appreciate if anyone here with experience of setting up a > calcium measurement system gives me some suggestion? > > > > Thank you, > > > > Yi > > > |
|
SUBSCRIBE CONFOCALMICROSCOPY Sara Smith |
|
|
In reply to this post by Boswell, Carl A - (cboswell)
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Carl, Thank you for your email. It is very good point. I will double check all of our objectives. Thanks again! Yi On Fri, Oct 21, 2011 at 12:55 PM, Boswell, Carl A - (cboswell) < [hidden email]> wrote: > Hi Sara, > An additional item not mentioned in Johannes' excellent reply is the > transmission limits of most lenses to UV wavelengths. If you are set on > using Fura2, seriously consider a quartz or other high-UV transmission > objective. Because so little 340nm light gets to the sample using "normal" > lenses, the fluorescent signal received from the 340 excitation remains > essentially unchanged with changes in intracellular calcium, rendering the > sensitivity of your measurements very low. > > Good luck, > C > > Carl A. Boswell > 520-954-7053 > > -----Original Message----- > From: Confocal Microscopy List [mailto:[hidden email]] > On Behalf Of SUBSCRIBE CONFOCALMICROSCOPY Sara Smith > Sent: Friday, October 21, 2011 8:12 AM > To: [hidden email] > Subject: Suggestion for setting up Calcium measurement system > > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Dear listers, > > We want to set up a calcium measurement system based on our old Nikon > microscpe (Eclipse TE300). It has a very nice environmental chamber and > equipped with Hamamatsu Digital Camera (C4742-95) and BioPoint filter wheel. > > We don't want something very fancy (like very high resolution or very high > speed acquisition) and really want to make a full use of what we have now > since we have a very limited budget. > Based on my research, we might need to buy a filter cube suitable for > calcium measurement (eg. Fura-2) and acquisition software with ratio imaging > application. Anything else do we need? > > I was looking at Hamamatsu AQUACOSMOS/Basic System. It is a very nice > system to do calcium measurement. However, it is quite pricy and sold as a > whole package. > > I would really appreciate if anyone here with experience of setting up a > calcium measurement system gives me some suggestion? > > Thank you, > > Yi > > |
|
Johannes Helm |
|
|
In reply to this post by SUBSCRIBE CONFOCALMICROSCOPY Sara Smith
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Dear Yi Zheng, > > For the measurement system, we want to have a least semi-quantitative > feature. > > We are planning to measure the calcium concentration by using ratio > metric dye. Ok. > > So the change of ratio of the intensity of two channels will present the > change of calcium concentration. At this stage, we might not need to > measure > the exactly calcium concentration in uM. Although is semi-quantitative, > we still want the time delay between the measurements of the two channels > to be as short as possible. > > > > The reason I listed Fura-2 as an example is I used to use it and the > microscope is equipped with a filter wheel, however we are open to any dye > that can fulfill this mission. Our microscope is a wide-field fluorescent > microscope not confocal. Is it still cheaper to go with an emission ratio > metric dye? If this is the case, we can reconsider to use indoes instead of > Fura-2. Since you already have a filter wheel attached to your microscope, you could use an excitation ratiometric dye. However, unless you would use an IMS style CLSM setup - you write that you are NOT planning to do this - excitation ratiometric measurements will un-avoidably result in time delays between the recordings of the fluorecenses of the Ca2+ bound and free dye molecules. Unless, of course, there is a technique avoiding this, which I am not aware of. For an emission ratiometric dye, you can buy equipment, which even in the wide field case will make it possible for you to avoid these time delays. Since you would like to attain a good time resolution, an emission ratiometric dye might, hence, be preferrable (provided that you, sooner or later, will have the budget to either buy a DV2 or possibly even a second camera and a DC2). An example for an emission ratiometric dye is Indo-1, although this dye has to be excited in the UV (excitation max. at approx. 350nm), what, as Carl Boswell has written, will be a problem on many microscope objectives. The subject is getting a little bit special and other group members possibly regard it as bothering. For the technical details, we might continue our discussion without Cc-ing to the confocal list (unless there are participants who still are interested, of course). Best wishes, Johannes > > > > As for the system set up, I was so clueless and your email helps me a > lot. > > It is very informative and guides me in the right direction. > > Thank you so much. > > I will read your paper and do more research about it. > > Thanks again! > > > > Yi > > > On Fri, Oct 21, 2011 at 11:50 AM, P. Johannes Helm > <[hidden email]>wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> ***** >> Good afternoon, >> an important decision - which most probably will be made by your budget - >> is whether you want to go for >> 1) >> a system that will be suitable for quantitative ratiometric measurements >> at minimum time delay between the measurements of the two channels, 2) >> a system that will be suitable for quantitative ratiometric measurements >> at which the time delay between the measurements of the two channels can >> be "long" - i.e. at least some seconds, >> 3) >> a system for qualitative measurements, only (at which there are some tricks to make even qualitative measurements a kind of >> semi-quantitative). >> I write this since you explicitely mentioned the dye substance "Fura-2" in >> your mail, which is an excitation ratiometric dye substance. >> Some comments: >> a) >> Before you consider ratiometric measurements at all, estimate your photon >> statistics. You will need quite a large number of photons to produce fully >> quantitative [Ca2+](x, y, z, t) images that are not drowned in noise, since the ratio imaging algorithm is strongly noise enhancing. Also, during your data analysis you should avoid using 8 bit or 12 bit values since truncation errors will have a terrible effect. >> Ref. for an analysis of that problem: >> Helm, Patwardhan, Manders, 1997, "A study of the precision of confocal, ratiometric, Fura-2-based [Ca2+]measurements", Cell Calcium >> 22(4):287-298 >> b) >> If you want to do ratiometric measurements, it will be cheaper to avoid excitation ratiometric dyes and go for an emission ratiometric dye, if possible, e.g. one of the Indoes. >> c) >> Whether excitation or emission ratiometric, you will - dependent on your >> measurement system - need either two filter cubes or one filter cube with >> two different filters for excitation or emission, which then will not be >> mounted in the cubes but somewhere else in the system (either on the illumination side for the excitation ratiometric system or on the emission >> side for the emission ratiometric system). >> d) >> Your budget might allow you to purchase one of these: >> http://www.photometrics.com/products/multichannel/dv2.php >> which will allow you to use your single camera as a two channel detector >> suitable even for fast- provided your rate of photons is large enough - emission ratiometric measurements. >> e) >> If ratiometry is beyond your financial horizon then you will most probably >> need to only purchase a suitable filtercube for your dye substance of choice. >> f) >> You may write your own data analysis software or check, e.g., ImageJ for >> Macros for your data analysis. Alternatively, you might inspect the market >> and find commercial data analysis software (possibly even combined with a >> suitable acquisition software). Since you have a Hamamatsu Camera, it might be a good idea to talk to your local Hamamatsu dealer about that. g) >> Definitely worth the investment even at a poor budget situation will be an >> electro mechanical shutter (e.g. the "Uniblitz" type by Vincent Associates, ref. www.uniblitz.com, or by JML, ref. >> http://www.jmloptical.com/pages/electronic-shutter-systems.aspx ) that can >> be triggered from your camera data acquisition software in order to minimize bleaching. >> Best wishes, >> Johannes >> -- >> P. Johannes Helm, M.Sc. PhD >> Seniorengineer >> CMBN >> University of Oslo >> Institute of Basic Medical Science >> Department of Anatomy >> Postboks 1105 - Blindern >> NO-0317 Oslo >> Voice: +47 228 51159 >> Fax: +47 228 51499 >> WWW: folk.uio.no/jhelm >> > ***** >> > To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> > ***** >> > >> > Dear listers, >> > >> > We want to set up a calcium measurement system based on our old Nikon >> microscpe (Eclipse TE300). It has a very nice environmental chamber and wheel. >> > >> > We don't want something very fancy (like very high resolution or very >> high >> > speed acquisition) and really want to make a full use of what we have >> now >> > since we have a very limited budget. >> > Based on my research, we might need to buy a filter cube suitable for >> calcium >> > measurement (eg. Fura-2) and acquisition software with ratio imaging >> application. Anything else do we need? >> > >> > I was looking at Hamamatsu AQUACOSMOS/Basic System. It is a very nice >> system to do calcium measurement. However, it is quite pricy and sold >> a >> > whole package. >> > >> > I would really appreciate if anyone here with experience of setting up >> a >> calcium measurement system gives me some suggestion? >> > >> > Thank you, >> > >> > Yi >> > > -- P. Johannes Helm, Dipl. Pys., tekn.lic., tekn.dr. Senioringeniør CMBN IMB Anatomi Postboks 1105 - Blindern kontor: Gaustad, Domus Medica, rom 0347 labor: Gaustad, Domus Medica, rom 0356 telefon: 51159 telefaks: 51499 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hello Everybody: It may be an off confocal question, but I was wondering if someone can point me to where I can get uniform sizes of carbon particles (5-20µm sizes). Thanks, -Prabhakar |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy ***** Hi Prabhakar, you can try Goodfellow. I ordered carbon particles long time ago there, but I don't know, if they have your special sizes. www.goodfellow.com - No commercial interest - Bye Wolfgang Am 20:59, schrieb B. Prabhakar Pandian: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > ***** > > Hello Everybody: > It may be an off confocal question, > but I was wondering if someone can point me to where I can get uniform > sizes of carbon particles (5-20µm sizes). > > Thanks, > > -Prabhakar -- Dr. Wolfgang Staroske Single Molecule Specialist Light Microscopy Facility Technische Universität Dresden Biotechnology Center Tatzberg 47/49 01307 Dresden, Germany Tel.: +49 (0) 351 463-40316 Fax.: +49 (0) 351 463-40342 E-Mail: [hidden email] Webpage: www.biotec.tu-dresden.de |
|
|
In reply to this post by SUBSCRIBE CONFOCALMICROSCOPY Sara Smith
Dear Sir,
I am using Fura2 dye for calcium staining with fluorescent microscope (Nikon). The filter which are required for this dye should have absorbance range (340-380nm) and emmission range (504-512nm). But due to unavailability of filter having both absorbance and emmission in any single filter,I used DAPI and FITC filter , but the result which we obtained have no any difference between Fura2 (dye) treated sample and control sample. Please suggest me how I solve this problem. Thanks Regards, Ums |