Erik Manders-2 |
![]() |
Dear all,
We would like to order a triple dichroic (405-440)/514/594 for our Spinning Disk system (purchased from Andor). However, since this would be a custom made dichroic the costs would be very high. Therefore, I would like to ask you: "Would you be interested in such filter?" (You are NOT obliged to buy one if you answer "yes") This dichroic will be suitable for the typical FRET-combinations such as CFP/YFP, YFP/mCherry(or mKate). Furthermore it is very suitable for swichable FP's (@ 405). If I find enough people who are interested in this dichroic, I could ask companies (Chroma, Semrock, Omega) to make one for us and sell it for a reasonable price. If you are interested, please directly reply to me. Later, I will report to the list how many replies I've got. Kind regards, Erik Manders |
![]() |
Hi Eric,
the one from the Chroma 89006 set will work (you may have up to 75% reflection for the 514 nm argon line, it is OK). As AR coating is required, ask Mike Stanley from Chroma, he is very helpful. Vitaly NCI-Frederick, 301-846-6575 Quoting Erik Manders <[hidden email]>: > Dear all, > > We would like to order a triple dichroic (405-440)/514/594 for our Spinning > Disk > system (purchased from Andor). However, since this would be a custom made > dichroic the costs would be very high. > > Therefore, I would like to ask you: "Would you be interested in such > filter?" > (You are NOT obliged to buy one if you answer "yes") > > This dichroic will be suitable for the typical FRET-combinations such as > CFP/YFP, YFP/mCherry(or mKate). Furthermore it is very suitable for swichable > > FP's (@ 405). > > If I find enough people who are interested in this dichroic, I could ask > companies (Chroma, Semrock, Omega) to make one for us and sell it for a > reasonable price. > > If you are interested, please directly reply to me. Later, I will report to > the list > how many replies I've got. > > Kind regards, Erik Manders > ------------------------------------------------- |
![]() ![]() |
Erik Manders-2 |
![]() |
In reply to this post by Erik Manders-2
Dear Vitaly,
Thanks for your reply. I think the Chroma 89006 is a set for a normal filter cube. Size of the dichroic that fits into the Yokogawa is much smaller and thinner. We have contacted Mike Stanley. He gave us spectra of other dichroics (fitting in the Yokogawa) but none of them is good for (405- 440)/514/594 . So, I think I really have to order a custom made dichrois. So, Please reply if you need one! Kind regards, Erik |
![]() ![]() |
Beat Ludin |
![]() |
Hi Erik
We have a (405-445)/515/561 filter set for the CSU series readily available, in case you are interested. It goes w/o saying that we have a commercial interest here. Best regards, Beat At 20:47 21-03-2009, you wrote: >Dear Vitaly, > >Thanks for your reply. I think the Chroma 89006 is a set for a normal filter >cube. Size of the dichroic that fits into the Yokogawa is much smaller and >thinner. We have contacted Mike Stanley. He gave us spectra of other >dichroics (fitting in the Yokogawa) but none of them is good for (405- >440)/514/594 . So, I think I really have to order a custom made dichrois. > >So, Please reply if you need one! > >Kind regards, Erik |
Free forum by Nabble | Edit this page |