Tyramide superboost kits quantitative?

classic Classic list List threaded Threaded
♦
♦
8 messages Options
Claire Brown Claire Brown
Reply | Threaded
Open this post in threaded view
♦
♦
|

Tyramide superboost kits quantitative?

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

I wonder if anyone knows if the new ThermoFisher Tyramide SuperBoost kits are quantitative?

 

https://www.thermofisher.com/order/catalog/product/B40915

 

 

I would assume that keeping the reaction conditions identical from sample to sample would be really important in stopping the HRP reaction after a fixed time but with such amplification is there any way relative signals could be quantified?

 

I would love to hear what people think.

 

Sincerely,

 

Claire

 

 

George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Tyramide superboost kits quantitative?

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Claire,

Depending on temperature, constant reaction time may be bad. Minimal issue is always run on a warm plate (that is perfectly uniform temperature ... there are slide thermometers for this), incubators or ovens.

I suspect the TSA reaction runs to completion usually, with one of these running out:

* active HRP enzyme molecules (self destructs with some probability for every H2O2 catalytic cycle)

* tyrosine "docking sites" ... both in cells/tissues, the antibody molecules, and the HRPs (an advantage for poly HRP)

* fluorescent tyramide (least likely?)

* H2O2 (also unlikely?)


I believe PerkinElmer's Opal kits for multiplexing deliberately strips the detection Ab-polyHRP, before the next cycle, but have not been able to get data from PerkinElmer on this (and did not need TSA while I worked in a research lab here in Houston).


Many IHC reagent companies sell HRP polymers conjugated to streptavidin or secondary antibodies. For example
STREPTAVIDIN POLY-HRP80 CONJUGATE (plus stabilizer diluent)
https://www.fitzgerald-fii.com/streptavidin-poly-hrp80-conjugate-65r-s118.html

Yes, you can replace components in the kit with other reagents ... and conversely, use components in the kit for other purposes. An interesting "head to head" is fluorescent tyramide vs DAB IHC, or whatever favorite HRP substrate detection reagent your friendly neighborhood pathologists like.

I'm a big fan of TSA ever  since Phil Moen of NEN (acquired by PerkinElmer) spoke about and demo'd it at a Cold Spring Harbor fluorescence in situ/immuno courses. Phil mentioned the diffusion radius of the activated tyramide radical is less than the diffraction limit of confocal microscopy (i.e. 100 nm radius, 200 nm diameter).

One potential limitation on quantitation: zero reactable tyrosines implies zero signal ... as I mentioned above, both the Ab and HRPs have tyrosines (some even on the surface if I recall correctly).

Big benefit: when done right, can dilute the primary antibody 100 fold (maybe more ... try it), saving money. In fact, failing to do so will result in massive background.

Automation: in principle, automating TSA should improve reproducibility (assuming constant temperature for the reaction, and either constant enzyme reaction time or "run to completion"). However, not paying attention to the cost of the reagents (or stability or lack thereof) could result in one well, $100 in reagents --> one slide).

One of my U Miami customers published a super-duper brightness CD4 and CD8 2 color T-cells (also CD4 and FoxP3) paper ... they killed off the HRP molecules (and before that, endogenous peroxidases) with a terrific product (both great name and works well): PeroxAbolish   http://biocare.net/product/peroxabolish/

https://www.ncbi.nlm.nih.gov/pubmed/21929847

Cell Transplant. 2012;21(1):113-25. doi: 10.3727/096368911X586747. Epub 2011 Sep 
16.

Quantitative in situ analysis of FoxP3+ T regulatory cells on transplant tissue
using laser scanning cytometry.

Takahashi H(1), Ruiz P, Ricordi C, Delacruz V, Miki A, Mita A, Misawa R, Barker
S, Burke GW, Tzakis AG, Ichii H.

Author information: 
(1)Miami Transplant Institute, University of Miami Leonard M. Miller School of
Medicine, Miami, FL, USA.

There is abundant evidence that immune cells infiltrating into a transplanted
organ play a critical role for destructive inflammatory or regulatory immune
reactions. Quantitative in situ analysis (i.e., in tissue sections) of immune
cells remains challenging due to a lack of objective methodology. Laser scanning 
cytometry (LSC) is an imaging-based methodology that performs quantitative
measurements on fluorescently and/ or chromatically stained tissue or cellular
specimens at a single-cell level. In this study, we have developed a novel
objective method for analysis of immune cells, including Foxp3(+) T regulatory
cells (Tregs), on formalin-fixed /paraffin-embedded (FFPE) transplant biopsy
sections using iCys® Research Imaging Cytometer. The development of multiple
immunofluorescent staining was established using FFPE human tonsil sample. The
CD4/CD8 ratio and the population of Tregs among CD4(+) cells were analyzed using 
iCys and compared with the results from conventional flow cytometry analysis
(FCM). Our multiple immunofluorescent staining techniques allow obtaining clear
staining on FFPE sections. The CD4/CD8 ratio analyzed by iCys was concordant with
those obtained by FCM. This method was also applicable for liver, small
intestine, kidney, pancreas, and heart transplant biopsy sections and provide an 
objective quantification of Tregs within the grafts.

DOI: 10.3727/096368911X586747 
PMCID: PMC3777543
PMID: 21929847  [PubMed - indexed for MEDLINE]
and my 2010 book chapter (Yuste 2010 CSHL Press chapter 15 - Imaging in Neurosciences) has a protocol (I would use Mol Probes/ThermoFisher or PerkinElmers ... can probably crosslink the Ab-HRPpolymer onto the cells/tissue instead of their strip).

I also want to acknowledge Molecular Probes - specifically Mike Janes and his team - for visiting Miami (several years ago) for a workshop featuring TSA and Bacmam2.0. Hopefully ThermoFisher still lets the Mol Probes folks out of the lab to do similar workshops.

Speculation (spectral-ation - I currently do not work in a lab, so not going to test this anytime soon -- maybe someone from Mol Probes can try it):

Single molecule localization staining ... or "rainbow single antigen counting by TSA" idea (this will be tedious to do by hand! Payoff: count everything):
=>I suggest using 35 mm imaging dish, so that wash steps can be very large volume (2 mL), and expensive reagent steps just the imaging area (7, 10, 14, 20, 28 mm diameter as appropriate for your expt).
1. incubate primary antibody at ~1/10th of saturation.
2. wash (extensively).
3. incubate secondary antibody-polyHRP.
4. wash extensively.
5. detect with Tyramide color #1.
6. kill HRP, i.e. with PeroxAbolish (http://biocare.net/product/peroxabolish/) AND I suggest gently crosslinking the 2ndAb-HRP-polymer into the specimen (see also Expansion Microscopy approaches).
7. OK, may want to image at this point, ideally with a way to refind the same place later.
*** Repeat 1-6 with different color, optionally also #7.
*. Image everything. See Valm et al https://www.ncbi.nlm.nih.gov/pubmed/27391327 for one example of high multiplexing (not tyramides) using all laser lines and detectors on a spectral confocal microscope.

I see in SuperBoost PDF only 7 colors (7 Alexa's)

https://tools.thermofisher.com/content/sfs/manuals/tyramide_superboost_kits_man.pdf

to get to 10 (ore more) could use biotin. A 1998 paper had do-it-yourself tyramide hapten synthesis
http://jhc.sagepub.com/content/46/6/771.full

Sure, could also do this single color, either add the signals, or photobleach. This could be useful on the MilliporeSigma/EMD Millipore/CellASIC ONIX microperfusion platform (since ONIX only has a few reservoirs and they seem to lack interest in making plates and manifolds for use with 2 or more controllers).

Another option for quantitation: wait for Garry P. Nolan (Stanford Univ.) to publish his immuofluorescence alternative to MIBI-ToF, or collaborate with him to use his MIBI-ToF or buy a Fluidigm imaging CyTOF.

enjoy,

George

On 10/13/2016 11:52 AM, Claire Brown, Dr. wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

I wonder if anyone knows if the new ThermoFisher Tyramide SuperBoost kits are quantitative?

 

https://www.thermofisher.com/order/catalog/product/B40915

 

 

I would assume that keeping the reaction conditions identical from sample to sample would be really important in stopping the HRP reaction after a fixed time but with such amplification is there any way relative signals could be quantified?

 

I would love to hear what people think.

 

Sincerely,

 

Claire

 

 


-- 


George McNamara, PhD
Houston, TX 77054
[hidden email]
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/
http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650

Sripad Ram-2 Sripad Ram-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Tyramide superboost kits quantitative?

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Hi George,
As always super informative. Is it just me or do others also think that its high time someone invents a device download all that info that's sitting in your head...?

Very interesting comment about the diffusion radius of the activated TSA radical being less than the diffraction limit. I think that is a very important parameter. Just wondering if you know a reference for that? 

Perkin Elmer recently showed some correlative fluorescence and EM data which suggests similar range of specificity for their TSA staining. 


Thanks!

Sripad Ram





On Thu, Oct 13, 2016 at 11:24 AM, George McNamara <[hidden email]> wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Claire,

Depending on temperature, constant reaction time may be bad. Minimal issue is always run on a warm plate (that is perfectly uniform temperature ... there are slide thermometers for this), incubators or ovens.

I suspect the TSA reaction runs to completion usually, with one of these running out:

* active HRP enzyme molecules (self destructs with some probability for every H2O2 catalytic cycle)

* tyrosine "docking sites" ... both in cells/tissues, the antibody molecules, and the HRPs (an advantage for poly HRP)

* fluorescent tyramide (least likely?)

* H2O2 (also unlikely?)


I believe PerkinElmer's Opal kits for multiplexing deliberately strips the detection Ab-polyHRP, before the next cycle, but have not been able to get data from PerkinElmer on this (and did not need TSA while I worked in a research lab here in Houston).


Many IHC reagent companies sell HRP polymers conjugated to streptavidin or secondary antibodies. For example
STREPTAVIDIN POLY-HRP80 CONJUGATE (plus stabilizer diluent)
https://www.fitzgerald-fii.com/streptavidin-poly-hrp80-conjugate-65r-s118.html

Yes, you can replace components in the kit with other reagents ... and conversely, use components in the kit for other purposes. An interesting "head to head" is fluorescent tyramide vs DAB IHC, or whatever favorite HRP substrate detection reagent your friendly neighborhood pathologists like.

I'm a big fan of TSA ever  since Phil Moen of NEN (acquired by PerkinElmer) spoke about and demo'd it at a Cold Spring Harbor fluorescence in situ/immuno courses. Phil mentioned the diffusion radius of the activated tyramide radical is less than the diffraction limit of confocal microscopy (i.e. 100 nm radius, 200 nm diameter).

One potential limitation on quantitation: zero reactable tyrosines implies zero signal ... as I mentioned above, both the Ab and HRPs have tyrosines (some even on the surface if I recall correctly).

Big benefit: when done right, can dilute the primary antibody 100 fold (maybe more ... try it), saving money. In fact, failing to do so will result in massive background.

Automation: in principle, automating TSA should improve reproducibility (assuming constant temperature for the reaction, and either constant enzyme reaction time or "run to completion"). However, not paying attention to the cost of the reagents (or stability or lack thereof) could result in one well, $100 in reagents --> one slide).

One of my U Miami customers published a super-duper brightness CD4 and CD8 2 color T-cells (also CD4 and FoxP3) paper ... they killed off the HRP molecules (and before that, endogenous peroxidases) with a terrific product (both great name and works well): PeroxAbolish   http://biocare.net/product/peroxabolish/

https://www.ncbi.nlm.nih.gov/pubmed/21929847

Cell Transplant. 2012;21(1):113-25. doi: 10.3727/096368911X586747. Epub 2011 Sep 
16.

Quantitative in situ analysis of FoxP3+ T regulatory cells on transplant tissue
using laser scanning cytometry.

Takahashi H(1), Ruiz P, Ricordi C, Delacruz V, Miki A, Mita A, Misawa R, Barker
S, Burke GW, Tzakis AG, Ichii H.

Author information: 
(1)Miami Transplant Institute, University of Miami Leonard M. Miller School of
Medicine, Miami, FL, USA.

There is abundant evidence that immune cells infiltrating into a transplanted
organ play a critical role for destructive inflammatory or regulatory immune
reactions. Quantitative in situ analysis (i.e., in tissue sections) of immune
cells remains challenging due to a lack of objective methodology. Laser scanning 
cytometry (LSC) is an imaging-based methodology that performs quantitative
measurements on fluorescently and/ or chromatically stained tissue or cellular
specimens at a single-cell level. In this study, we have developed a novel
objective method for analysis of immune cells, including Foxp3(+) T regulatory
cells (Tregs), on formalin-fixed /paraffin-embedded (FFPE) transplant biopsy
sections using iCys® Research Imaging Cytometer. The development of multiple
immunofluorescent staining was established using FFPE human tonsil sample. The
CD4/CD8 ratio and the population of Tregs among CD4(+) cells were analyzed using 
iCys and compared with the results from conventional flow cytometry analysis
(FCM). Our multiple immunofluorescent staining techniques allow obtaining clear
staining on FFPE sections. The CD4/CD8 ratio analyzed by iCys was concordant with
those obtained by FCM. This method was also applicable for liver, small
intestine, kidney, pancreas, and heart transplant biopsy sections and provide an 
objective quantification of Tregs within the grafts.

DOI: 10.3727/096368911X586747 
PMCID: PMC3777543
PMID: 21929847  [PubMed - indexed for MEDLINE]
and my 2010 book chapter (Yuste 2010 CSHL Press chapter 15 - Imaging in Neurosciences) has a protocol (I would use Mol Probes/ThermoFisher or PerkinElmers ... can probably crosslink the Ab-HRPpolymer onto the cells/tissue instead of their strip).

I also want to acknowledge Molecular Probes - specifically Mike Janes and his team - for visiting Miami (several years ago) for a workshop featuring TSA and Bacmam2.0. Hopefully ThermoFisher still lets the Mol Probes folks out of the lab to do similar workshops.

Speculation (spectral-ation - I currently do not work in a lab, so not going to test this anytime soon -- maybe someone from Mol Probes can try it):

Single molecule localization staining ... or "rainbow single antigen counting by TSA" idea (this will be tedious to do by hand! Payoff: count everything):
=>I suggest using 35 mm imaging dish, so that wash steps can be very large volume (2 mL), and expensive reagent steps just the imaging area (7, 10, 14, 20, 28 mm diameter as appropriate for your expt).
1. incubate primary antibody at ~1/10th of saturation.
2. wash (extensively).
3. incubate secondary antibody-polyHRP.
4. wash extensively.
5. detect with Tyramide color #1.
6. kill HRP, i.e. with PeroxAbolish (http://biocare.net/product/peroxabolish/) AND I suggest gently crosslinking the 2ndAb-HRP-polymer into the specimen (see also Expansion Microscopy approaches).
7. OK, may want to image at this point, ideally with a way to refind the same place later.
*** Repeat 1-6 with different color, optionally also #7.
*. Image everything. See Valm et al https://www.ncbi.nlm.nih.gov/pubmed/27391327 for one example of high multiplexing (not tyramides) using all laser lines and detectors on a spectral confocal microscope.

I see in SuperBoost PDF only 7 colors (7 Alexa's)

https://tools.thermofisher.com/content/sfs/manuals/tyramide_superboost_kits_man.pdf

to get to 10 (ore more) could use biotin. A 1998 paper had do-it-yourself tyramide hapten synthesis
http://jhc.sagepub.com/content/46/6/771.full

Sure, could also do this single color, either add the signals, or photobleach. This could be useful on the MilliporeSigma/EMD Millipore/CellASIC ONIX microperfusion platform (since ONIX only has a few reservoirs and they seem to lack interest in making plates and manifolds for use with 2 or more controllers).

Another option for quantitation: wait for Garry P. Nolan (Stanford Univ.) to publish his immuofluorescence alternative to MIBI-ToF, or collaborate with him to use his MIBI-ToF or buy a Fluidigm imaging CyTOF.

enjoy,

George


On 10/13/2016 11:52 AM, Claire Brown, Dr. wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

I wonder if anyone knows if the new ThermoFisher Tyramide SuperBoost kits are quantitative?

 

https://www.thermofisher.com/order/catalog/product/B40915

 

 

I would assume that keeping the reaction conditions identical from sample to sample would be really important in stopping the HRP reaction after a fixed time but with such amplification is there any way relative signals could be quantified?

 

I would love to hear what people think.

 

Sincerely,

 

Claire

 

 


-- 


George McNamara, PhD
Houston, TX 77054
[hidden email]
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/
http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650


George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Tyramide superboost kits quantitative?

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Sripad,

Phillip Moen is the best I can do for TSA products per se. His linkedin profile is

https://www.linkedin.com/in/phillip-moen-2642184

Tom Deerink, Mark Ellisman, roger Tsien et al have correlative LM-EM for diaminobenzidine from eosin, miniSOG, etc. The DAB monomer is al little smaller than fluorophore-tyramide, likely similar radicals either way.

Now with super-resolution, say collagen 2D matrix, detect with limiting dilution anti-collagen, detect that with HRP- or HRP-polymer antibody or streptavidin, and detect with Alexa Fluor 647-tyramide, add the latest best super-res mounting medium, acquire a couple of images and publish.

Molecular Probes (Thermo Fisher) licensed TSA from PerkinElmer, so the fundamentals are the same. Details in the reagent kits probably do not matter much and will be mostly interchangable.

My brain dump on microscopy through ~2005 is in the MPMicro download at

https://works.bepress.com/gmcnamara/2

FP table a bit out of date, see http://www.geomcnamara.com/fluorescent-proteins-photophysics-data

I haven't looked to see if I had FP biosensors, current table of very basic info is at http://www.geomcnamara.com/fluorescent-biosensors 

FP Biosensors table is currently lacking key references -- these are:

  http://biosensor.dpb.carnegiescience.edu/

  https://www.ncbi.nlm.nih.gov/pubmed/21456512   and their supplemental file at http://pubs.acs.org/doi/suppl/10.1021/cr100002u


At some point I'll post the fluorophores table from my Current Protocols unit, but that is a limited list, and not much different than in the PubSpectra XLSX file inside the ZIP download at

https://works.bepress.com/gmcnamara/9/


George


On 10/13/2016 6:20 PM, S Ram wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Hi George,
As always super informative. Is it just me or do others also think that its high time someone invents a device download all that info that's sitting in your head...?

Very interesting comment about the diffusion radius of the activated TSA radical being less than the diffraction limit. I think that is a very important parameter. Just wondering if you know a reference for that? 

Perkin Elmer recently showed some correlative fluorescence and EM data which suggests similar range of specificity for their TSA staining. 


Thanks!

Sripad Ram





On Thu, Oct 13, 2016 at 11:24 AM, George McNamara <[hidden email]> wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Claire,

Depending on temperature, constant reaction time may be bad. Minimal issue is always run on a warm plate (that is perfectly uniform temperature ... there are slide thermometers for this), incubators or ovens.

I suspect the TSA reaction runs to completion usually, with one of these running out:

* active HRP enzyme molecules (self destructs with some probability for every H2O2 catalytic cycle)

* tyrosine "docking sites" ... both in cells/tissues, the antibody molecules, and the HRPs (an advantage for poly HRP)

* fluorescent tyramide (least likely?)

* H2O2 (also unlikely?)


I believe PerkinElmer's Opal kits for multiplexing deliberately strips the detection Ab-polyHRP, before the next cycle, but have not been able to get data from PerkinElmer on this (and did not need TSA while I worked in a research lab here in Houston).


Many IHC reagent companies sell HRP polymers conjugated to streptavidin or secondary antibodies. For example
STREPTAVIDIN POLY-HRP80 CONJUGATE (plus stabilizer diluent)
https://www.fitzgerald-fii.com/streptavidin-poly-hrp80-conjugate-65r-s118.html

Yes, you can replace components in the kit with other reagents ... and conversely, use components in the kit for other purposes. An interesting "head to head" is fluorescent tyramide vs DAB IHC, or whatever favorite HRP substrate detection reagent your friendly neighborhood pathologists like.

I'm a big fan of TSA ever  since Phil Moen of NEN (acquired by PerkinElmer) spoke about and demo'd it at a Cold Spring Harbor fluorescence in situ/immuno courses. Phil mentioned the diffusion radius of the activated tyramide radical is less than the diffraction limit of confocal microscopy (i.e. 100 nm radius, 200 nm diameter).

One potential limitation on quantitation: zero reactable tyrosines implies zero signal ... as I mentioned above, both the Ab and HRPs have tyrosines (some even on the surface if I recall correctly).

Big benefit: when done right, can dilute the primary antibody 100 fold (maybe more ... try it), saving money. In fact, failing to do so will result in massive background.

Automation: in principle, automating TSA should improve reproducibility (assuming constant temperature for the reaction, and either constant enzyme reaction time or "run to completion"). However, not paying attention to the cost of the reagents (or stability or lack thereof) could result in one well, $100 in reagents --> one slide).

One of my U Miami customers published a super-duper brightness CD4 and CD8 2 color T-cells (also CD4 and FoxP3) paper ... they killed off the HRP molecules (and before that, endogenous peroxidases) with a terrific product (both great name and works well): PeroxAbolish   http://biocare.net/product/peroxabolish/

https://www.ncbi.nlm.nih.gov/pubmed/21929847

Cell Transplant. 2012;21(1):113-25. doi: 10.3727/096368911X586747. Epub 2011 Sep 
16.

Quantitative in situ analysis of FoxP3+ T regulatory cells on transplant tissue
using laser scanning cytometry.

Takahashi H(1), Ruiz P, Ricordi C, Delacruz V, Miki A, Mita A, Misawa R, Barker
S, Burke GW, Tzakis AG, Ichii H.

Author information: 
(1)Miami Transplant Institute, University of Miami Leonard M. Miller School of
Medicine, Miami, FL, USA.

There is abundant evidence that immune cells infiltrating into a transplanted
organ play a critical role for destructive inflammatory or regulatory immune
reactions. Quantitative in situ analysis (i.e., in tissue sections) of immune
cells remains challenging due to a lack of objective methodology. Laser scanning 
cytometry (LSC) is an imaging-based methodology that performs quantitative
measurements on fluorescently and/ or chromatically stained tissue or cellular
specimens at a single-cell level. In this study, we have developed a novel
objective method for analysis of immune cells, including Foxp3(+) T regulatory
cells (Tregs), on formalin-fixed /paraffin-embedded (FFPE) transplant biopsy
sections using iCys® Research Imaging Cytometer. The development of multiple
immunofluorescent staining was established using FFPE human tonsil sample. The
CD4/CD8 ratio and the population of Tregs among CD4(+) cells were analyzed using 
iCys and compared with the results from conventional flow cytometry analysis
(FCM). Our multiple immunofluorescent staining techniques allow obtaining clear
staining on FFPE sections. The CD4/CD8 ratio analyzed by iCys was concordant with
those obtained by FCM. This method was also applicable for liver, small
intestine, kidney, pancreas, and heart transplant biopsy sections and provide an 
objective quantification of Tregs within the grafts.

DOI: 10.3727/096368911X586747 
PMCID: PMC3777543
PMID: 21929847  [PubMed - indexed for MEDLINE]
and my 2010 book chapter (Yuste 2010 CSHL Press chapter 15 - Imaging in Neurosciences) has a protocol (I would use Mol Probes/ThermoFisher or PerkinElmers ... can probably crosslink the Ab-HRPpolymer onto the cells/tissue instead of their strip).

I also want to acknowledge Molecular Probes - specifically Mike Janes and his team - for visiting Miami (several years ago) for a workshop featuring TSA and Bacmam2.0. Hopefully ThermoFisher still lets the Mol Probes folks out of the lab to do similar workshops.

Speculation (spectral-ation - I currently do not work in a lab, so not going to test this anytime soon -- maybe someone from Mol Probes can try it):

Single molecule localization staining ... or "rainbow single antigen counting by TSA" idea (this will be tedious to do by hand! Payoff: count everything):
=>I suggest using 35 mm imaging dish, so that wash steps can be very large volume (2 mL), and expensive reagent steps just the imaging area (7, 10, 14, 20, 28 mm diameter as appropriate for your expt).
1. incubate primary antibody at ~1/10th of saturation.
2. wash (extensively).
3. incubate secondary antibody-polyHRP.
4. wash extensively.
5. detect with Tyramide color #1.
6. kill HRP, i.e. with PeroxAbolish (http://biocare.net/product/peroxabolish/) AND I suggest gently crosslinking the 2ndAb-HRP-polymer into the specimen (see also Expansion Microscopy approaches).
7. OK, may want to image at this point, ideally with a way to refind the same place later.
*** Repeat 1-6 with different color, optionally also #7.
*. Image everything. See Valm et al https://www.ncbi.nlm.nih.gov/pubmed/27391327 for one example of high multiplexing (not tyramides) using all laser lines and detectors on a spectral confocal microscope.

I see in SuperBoost PDF only 7 colors (7 Alexa's)

https://tools.thermofisher.com/content/sfs/manuals/tyramide_superboost_kits_man.pdf

to get to 10 (ore more) could use biotin. A 1998 paper had do-it-yourself tyramide hapten synthesis
http://jhc.sagepub.com/content/46/6/771.full

Sure, could also do this single color, either add the signals, or photobleach. This could be useful on the MilliporeSigma/EMD Millipore/CellASIC ONIX microperfusion platform (since ONIX only has a few reservoirs and they seem to lack interest in making plates and manifolds for use with 2 or more controllers).

Another option for quantitation: wait for Garry P. Nolan (Stanford Univ.) to publish his immuofluorescence alternative to MIBI-ToF, or collaborate with him to use his MIBI-ToF or buy a Fluidigm imaging CyTOF.

enjoy,

George


On 10/13/2016 11:52 AM, Claire Brown, Dr. wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

I wonder if anyone knows if the new ThermoFisher Tyramide SuperBoost kits are quantitative?

 

https://www.thermofisher.com/order/catalog/product/B40915

 

 

I would assume that keeping the reaction conditions identical from sample to sample would be really important in stopping the HRP reaction after a fixed time but with such amplification is there any way relative signals could be quantified?

 

I would love to hear what people think.

 

Sincerely,

 

Claire

 

 


-- 


George McNamara, PhD
Houston, TX 77054
[hidden email]
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/
http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650

-- 


George McNamara, PhD
Houston, TX 77054
[hidden email]
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/
http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650

George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Fluorophores Photophysics Data

In reply to this post by Sripad Ram-2
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Dear Confocal Listserv readers,

If any of your favorite fluorescent dyes (small molecules, not fluorescent proteins) are not in the draft table below, please send them to me in the same column format, and I can add them to my draft. The table should appear in my Current Protocols chapter, and sooner online at http://www.geomcnamara.com/data which already has a fluorescent proteins data table (not photo$able FPs) and very basic list of FP biosensors. You are free to repost: data are facts, facts are not copyrightable.

Enjoy,

George

Table 4.4.2 Fluorophores Photophysics Data

Fluorophore

Abs or Ex max (nm)

Em max (nm)

Extinction Coefficient

Quantum Yield

Brightness Index

Acridine orange

271

520

27,000

0.20

5

Alexa Fluor 430

431

541

16,000

0.55

9

Alexa Fluor 488

495

519

71,000

0.94

67

Alexa Fluor 532

532

553

81,000

0.80

65

Alexa Fluor 546

556

573

104,000

0.96

100

Alexa Fluor 568

578

603

91,300

0.75

69

Alexa Fluor 594

590

617

73,000

0.64

47

Allophycocyanin (APC)

650

660

700,000

0.68

476

Atto 390

390

479

24,000

0.90

22

Atto 425

436

484

45,000

0.65

29

Atto 430LS

433

547

32,000

0.65

21

Atto 465

453

508

75,000

0.75

56

Atto 490LS

496

661

40,000

0.30

12

Atto 520

525

547

105,000

0.95

100

Atto 532

534

560

115,000

0.90

104

Atto 565

566

590

120,000

0.97

116

Atto 590

598

634

120,000

0.90

108

Atto 610

616

646

150,000

0.70

105

Atto 620

620

641

120,000

0.50

60

Atto 635

637

660

120,000

0.45

54

Atto 655

655

680

125,000

0.50

63

Atto 680

675

699

125,000

0.40

50

ATTO-Dino 1 (dsDNA)

490

531

179,000

0.70

125

ATTO-Dino 2 (dsDNA)

506

535

162,000

0.70

113

Bacteriochlorin (“BC”)

713

717

120,000

0.11

13

Bacteriochlorin BC-T2-T12

737

744

120,000

0.14

17

Bacteriochlorin BC-Ph3-Ph13

736

742

120,000

0.12

14

Bacteriochlorin BC-T2-OMe5-T12

732

739

120,000

0.18

22

Bacteriochlorin BC-V3-V13

750

756

120,000

0.10

12

Bacteriochlorin BC-PE3-PE13

763

768

120,000

0.15

18

Bacteriochlorin BC-A3-A13

768

774

120,000

0.09

11

Bacteriochlorin BC-F3-F13

771

777

120,000

0.07

8

BODIPY 507/545

513

549

82,800

0.73

60

BODIPY FL

504

510

70,000

0.90

63

BODIPY TR

588

616

68,000

0.84

57

B-phycoerythrin (B-PE)

545

575

2,410,000

0.98

2,362

Calcein

494

516

81,000

0.78

63

Cascade Blue

378

423

26,000

0.54

14

Chromeo 488

488

517

73,000

0.27

20

Chromeo 494

494

628

55,000

0.15

8

Chromeo 505

505

526

70,000

0.30

21

Chromeo 546

545

561

98,800

0.15

15

Chromeo 642

642

660

180,000

0.21

38

Chromeo P429 Py-Dye

429

536

75,000

0.10

8

Chromeo P503 Py-Dye

503

600

24,000

0.50

12

Chromeo P540 Py-Dye

533

627

50,000

0.20

10

Chromeo P543 Py-Dye

543

590

57,000

0.15

9

Coumarin 6

456

500

54,000

0.78

42

Cresyl violet perchlorate

603

622

83,000

0.54

45

Cy3

552

570

150,000

0.15

23

Cy3B

552

570

130,000

0.67

87

Cy5

649

670

250,000

0.28

70

Cy5.5

675

694

250,000

0.23

58

Cy7

755

778

250,000

0.28

70

DAPI (in DMSO)

353

465

27,000

0.58

16

DAPI (in H2O)

344

487

27,000

0.04

1

DsRed

558

583

75,000

0.70

52

Eosin Y

525

543

112,000

0.67

75

EYFP

514

527

84,000

0.61

51

Fluorescein

490

514

90,000

0.92

83

Fluorescein FH3+ (pH <3)

437

-

53,000

-

-

Fluorescein FH2 (pH 4)

434

-

11,000

-

-

Fluorescein FH- (pH 5.3)

472

515

29,000

0.37

11

Fluorescein F2- (pH >8)

490

515

76,900

0.92

71

FM 1-43

479

598

40,000

0.30

12

Fura-2, Ca++ free

363

512

28,000

0.23

6

Fura-2, Ca++ saturated

335

505

34,000

0.49

17

Fura-2, Mn++ saturated

335

-

-

0

0

Fura-2, Zn++ saturated

345

505

34,000

0.69

24

Hoechst 33258 (in DMF)

354

486

46,000

0.35

16

Hoechst 33258 (in H2O)

345

507

46,000

0.03

2

Indo-1, Ca++ free

346

475

33,000

0.38

13

Indo-1, Ca++ saturated

330

401

33,000

0.56

18

IRDye38

778

806

179,000

0.35

62

IRDye40

768

788

140,000

0.38

53

IRDye700

681

712

170,000

0.48

81

IRDye78

768

796

220,000

0.31

68

IRDye80

767

791

250,000

0.21

53

IRDye800

787

812

275,000

0.15

41

JOE

520

548

73,000

0.60

44

Lucifer Yellow CH

230

542

24,200

0.21

5

merocyanine 540

559

579

138,000

0.39

54

neo-Cy5 (DMSO)

656

675

195,000

0.25

49

NIR1

761

796

268,000

0.23

62

NIR2

662

684

250,000

0.34

85

NIR3

750

777

275,000

0.28

77

NIR4

650

671

260,000

0.43

1112

Oregon Green 488

496

516

76,000

0.90

68

Oregon Green 514

506

526

88,000

0.96

85

Oyster® - 645 (ethanol)

651

669

250,000

0.40

100

Oyster® - 656 (ethanol)

665

684

220,000

0.50

11

Pacific Blue

400

447

29,500

0.55

16

Perylene

253

435

38,500

0.94

36

Phenylalanine

222

279

195

0.02

<0.1

POPOP

256

407

47,000

0.93

44

Quinine sulfate (in 0.5M H2SO4)

256

451

5,700

0.55

3

Rhodamine 110

496

520

80,000

0.89

71

Rhodamine 6G

530

552

116,000

0.95

110

Rhodamine B

543

565

106,000

0.70

74

Riboflavin

262

531

34,800

0.37

13

Rose bengal

559

571

90,400

0.11

10

R-Phycoerythrin (R-PE)

480

578

1,960,000

0.68

1,333

SNIR1

666

695

218,000

0.24

52

SNIR3

667

697

245,000

0.24

59

Star 440 SXP

436

515

22,700

0.68

15

Star 470 SXP

472

624

29,000

0.12

4

Star 488

503

524

64,500

0.89

57

Star 512

511

530

84,000

0.82

69

Star 520SXP

515

612

60,000

0.05

3

Star 580

587

607

72,000

0.90

65

Star 600

604

627

43,500

0.73

32

Star 635

639

654

63,000

0.51

32

Star 635P

635

651

125,000

0.92

115

Star Red

638

655

212,000

0.90

191

Sulforhodamine 101

576

591

139,000

0.90

125

Texas Red

586

605

108,000

0.77

83

Texas Red-X

583

603

116,000

0.90

104

TMR

540

565

95,000

0.68

65

Trp

287

348

6,000

0.31

2

Tyr

275

303

1,500

0.21

0.3

Brilliant Violets, Brilliant Ultraviolets, Brilliant Blue

Brilliant Violet BV421

405

421

2,500,000

0.65

1,625

Brilliant Violet BV510

405

510

577,000

0.44

254

Brilliant Violet BV570

405

570

2,300,000

0.08

184

Brilliant Violet BV605

405

603

2,400,000

0.29

696

Brilliant Violet BV650

405

645

2,500,000

0.17

425

Brilliant Violet BV711

405

711

2,800,000

0.15

420

Brilliant Violet BV785

405

785

2,500,000

0.04

100

Brilliant Ultraviolet BUV395

348

395

 

 

 

Brilliant Ultraviolet BUV496

348

496

 

 

 

Brilliant Ultraviolet BUV563

348

563

 

 

 

Brilliant Ultraviolet BUV661

348

661

 

 

 

Brilliant Ultraviolet BUV737

348

737

 

 

 

Brilliant Ultraviolet BUV805

348

805

 

 

 

Brilliant Blue BB515

490

515

 

 

 

Fluorescent nanocrystal quantum dots

Product

Em Peak

Em FWHM

Extinction

Coefficient

Quantum

Yield

Brightness

Index

QD525

525

≤32

320,000

0.60

192

QD565

565

≤34

1,100,000

0.40

440

QD585

585

≤34

2,200,000

0.40

880

QD605

605

≤27

2,400,000

0.40

960

QD625

625

 

 

 

>2,300

QD655

655

≤34

5,700,000

0.40

2,280

QD705

705

wide

 

 

 

eVolve 605

605

 

 

 

 

eVolve 655

655

 

 

 

 

Brightness Index = Extinction coefficient * Quantum Yield / 1000. Table features fluorophores for which all four major photophysical parameters, absorption (excitation) maximum, fluorescence emission maximum extinction coefficient (M-1 cm-1) and fluorescence quantum yield, are available. QD625 values details have not been published, but is brighter than QD605 or QD655.

Antibody, streptavidin, tetramers/pentamers/dextramers, and other macromolecular labeling have “degree of labeling’ (doL) issues. As a simplistic rule of thumb, one fluorophore can be conjugated to the surface of a protein for every 25,000 dalton molecular weight. An IgG molecule is ~155,000 dalton, so ~6 fluorophores can be conjugated without too much quenching or causing the protein to crash out of solution. With random labeling of lysines, such as isothiocyantate chemistry (FITC, TRITC), there is a risk of the reactive dye making a covalent bond in an Fab binding site, which would occlude binding to the target antigen epitope. If both Fabs were occluded, this dye-IgG would not bind antigenic epitope at all, but could bind non-specifically, increasing background. If there are 50 reactive lysines on the surface of an IgG molecule, and degree of labeling is 6 (+/- some variability), the population of antibody molecules applied to a specimen is going to be very heterogeneous: a small reaction may not have any identical antibody molecules. DoL of different batches of antibodies are unlikely to be identical. Isotype controls, whether a mouse IgG1 monoclonal antibody (defined specificity different than your interest), or polyclonal affinity purified from sera, are at best imperfect controls. Immunoglobulins also vary in glycosylation, disulfide bonding, and single nucleotide polymorphisms – including some that result in amino acid substitutions that affect Fc Receptor binding affinity. A recent (2015) trend has been for vendors, ex. Miltenyi Biotec REAfinity, to use the same “backbone’ for their entire product line antibodies.

The Fluorophore dyes section of the table is abridged from a 4700+ entry data table Excel file the authors posted on the internet,

https://works.bepress.com/gmcnamara/9  = PubSpectra.

the website download also contains 400+ entry fluorescent proteins data table Excel file. Spectra for many of these dyes, and commercial filters and light sources, are available through an interactive website at

http://www.spectra.arizona.edu/

QD### data from Quantum Dot Corp (acquired by Molecular Probes / Thermo Fisher). eVolve 605 and eVolve 655 are Cd containing quantum dots from Affymetrix / eBiosciences (acquired by Thermo Fisher). Bacteriochlorins (BC’s) have small Stokes shifts (Emission maximum – Excitation maximum). They have almost as large extinction coefficients in the ultraviolet, 350-390 nm range, so for practical use have a huge, practical Stokes shifts on the order of 400 nm (they also have an additional excitation peak in the 480-550 nm range. Taniguchi et al (2008) include a graph comparing the full width half maxima (FWHM) emission spectra of Cyanine dyes (~50 nm), near infrared quantum dots (66 nm QD705, 74 nm QD800), BC’s (~20 nm), and ‘expanded’ porphyrins (>120 nm). We suggest with the right excitation light sources (355 nm laser or ~360 and ~390 nm LEDs), and emission separation (filter wheel one camera, cascading filters to multiple cameras or PMTs), that four or more BC’s could be 700-800 nm emission range. If a palette of five BC’s could be closely conjugated to Brilliant Blue BB515, excitation peak 488 nm, extremely efficient energy transfer in the visible absorption peaks (~500-550 nm) could enable extreme multiplexing. Additional multiplexing may be gained by 355 nm excitation of BUV395 to five BC’s. Our thanks to Dr. Bruce Pitner, NIRvana Sciences, for discussions extinction coefficient estimates with respect to NIRvana dyes and Taniguchi et al (2008) and Taniguchi and Lindsey (2016).. Additional dye data from Sednev et al (2015, BioLegend, 2016, Brilliant Violets, http://www.biolegend.com/brilliantviolet 

You can look for updates at at www.GeoMcNamara.com.


On 10/13/2016 6:20 PM, S Ram wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Hi George,
As always super informative. Is it just me or do others also think that its high time someone invents a device download all that info that's sitting in your head...?

Very interesting comment about the diffusion radius of the activated TSA radical being less than the diffraction limit. I think that is a very important parameter. Just wondering if you know a reference for that? 

Perkin Elmer recently showed some correlative fluorescence and EM data which suggests similar range of specificity for their TSA staining. 


Thanks!

Sripad Ram





On Thu, Oct 13, 2016 at 11:24 AM, George McNamara <[hidden email]> wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Claire,

Depending on temperature, constant reaction time may be bad. Minimal issue is always run on a warm plate (that is perfectly uniform temperature ... there are slide thermometers for this), incubators or ovens.

I suspect the TSA reaction runs to completion usually, with one of these running out:

* active HRP enzyme molecules (self destructs with some probability for every H2O2 catalytic cycle)

* tyrosine "docking sites" ... both in cells/tissues, the antibody molecules, and the HRPs (an advantage for poly HRP)

* fluorescent tyramide (least likely?)

* H2O2 (also unlikely?)


I believe PerkinElmer's Opal kits for multiplexing deliberately strips the detection Ab-polyHRP, before the next cycle, but have not been able to get data from PerkinElmer on this (and did not need TSA while I worked in a research lab here in Houston).


Many IHC reagent companies sell HRP polymers conjugated to streptavidin or secondary antibodies. For example
STREPTAVIDIN POLY-HRP80 CONJUGATE (plus stabilizer diluent)
https://www.fitzgerald-fii.com/streptavidin-poly-hrp80-conjugate-65r-s118.html

Yes, you can replace components in the kit with other reagents ... and conversely, use components in the kit for other purposes. An interesting "head to head" is fluorescent tyramide vs DAB IHC, or whatever favorite HRP substrate detection reagent your friendly neighborhood pathologists like.

I'm a big fan of TSA ever  since Phil Moen of NEN (acquired by PerkinElmer) spoke about and demo'd it at a Cold Spring Harbor fluorescence in situ/immuno courses. Phil mentioned the diffusion radius of the activated tyramide radical is less than the diffraction limit of confocal microscopy (i.e. 100 nm radius, 200 nm diameter).

One potential limitation on quantitation: zero reactable tyrosines implies zero signal ... as I mentioned above, both the Ab and HRPs have tyrosines (some even on the surface if I recall correctly).

Big benefit: when done right, can dilute the primary antibody 100 fold (maybe more ... try it), saving money. In fact, failing to do so will result in massive background.

Automation: in principle, automating TSA should improve reproducibility (assuming constant temperature for the reaction, and either constant enzyme reaction time or "run to completion"). However, not paying attention to the cost of the reagents (or stability or lack thereof) could result in one well, $100 in reagents --> one slide).

One of my U Miami customers published a super-duper brightness CD4 and CD8 2 color T-cells (also CD4 and FoxP3) paper ... they killed off the HRP molecules (and before that, endogenous peroxidases) with a terrific product (both great name and works well): PeroxAbolish   http://biocare.net/product/peroxabolish/

https://www.ncbi.nlm.nih.gov/pubmed/21929847

Cell Transplant. 2012;21(1):113-25. doi: 10.3727/096368911X586747. Epub 2011 Sep 
16.

Quantitative in situ analysis of FoxP3+ T regulatory cells on transplant tissue
using laser scanning cytometry.

Takahashi H(1), Ruiz P, Ricordi C, Delacruz V, Miki A, Mita A, Misawa R, Barker
S, Burke GW, Tzakis AG, Ichii H.

Author information: 
(1)Miami Transplant Institute, University of Miami Leonard M. Miller School of
Medicine, Miami, FL, USA.

There is abundant evidence that immune cells infiltrating into a transplanted
organ play a critical role for destructive inflammatory or regulatory immune
reactions. Quantitative in situ analysis (i.e., in tissue sections) of immune
cells remains challenging due to a lack of objective methodology. Laser scanning 
cytometry (LSC) is an imaging-based methodology that performs quantitative
measurements on fluorescently and/ or chromatically stained tissue or cellular
specimens at a single-cell level. In this study, we have developed a novel
objective method for analysis of immune cells, including Foxp3(+) T regulatory
cells (Tregs), on formalin-fixed /paraffin-embedded (FFPE) transplant biopsy
sections using iCys® Research Imaging Cytometer. The development of multiple
immunofluorescent staining was established using FFPE human tonsil sample. The
CD4/CD8 ratio and the population of Tregs among CD4(+) cells were analyzed using 
iCys and compared with the results from conventional flow cytometry analysis
(FCM). Our multiple immunofluorescent staining techniques allow obtaining clear
staining on FFPE sections. The CD4/CD8 ratio analyzed by iCys was concordant with
those obtained by FCM. This method was also applicable for liver, small
intestine, kidney, pancreas, and heart transplant biopsy sections and provide an 
objective quantification of Tregs within the grafts.

DOI: 10.3727/096368911X586747 
PMCID: PMC3777543
PMID: 21929847  [PubMed - indexed for MEDLINE]
and my 2010 book chapter (Yuste 2010 CSHL Press chapter 15 - Imaging in Neurosciences) has a protocol (I would use Mol Probes/ThermoFisher or PerkinElmers ... can probably crosslink the Ab-HRPpolymer onto the cells/tissue instead of their strip).

I also want to acknowledge Molecular Probes - specifically Mike Janes and his team - for visiting Miami (several years ago) for a workshop featuring TSA and Bacmam2.0. Hopefully ThermoFisher still lets the Mol Probes folks out of the lab to do similar workshops.

Speculation (spectral-ation - I currently do not work in a lab, so not going to test this anytime soon -- maybe someone from Mol Probes can try it):

Single molecule localization staining ... or "rainbow single antigen counting by TSA" idea (this will be tedious to do by hand! Payoff: count everything):
=>I suggest using 35 mm imaging dish, so that wash steps can be very large volume (2 mL), and expensive reagent steps just the imaging area (7, 10, 14, 20, 28 mm diameter as appropriate for your expt).
1. incubate primary antibody at ~1/10th of saturation.
2. wash (extensively).
3. incubate secondary antibody-polyHRP.
4. wash extensively.
5. detect with Tyramide color #1.
6. kill HRP, i.e. with PeroxAbolish (http://biocare.net/product/peroxabolish/) AND I suggest gently crosslinking the 2ndAb-HRP-polymer into the specimen (see also Expansion Microscopy approaches).
7. OK, may want to image at this point, ideally with a way to refind the same place later.
*** Repeat 1-6 with different color, optionally also #7.
*. Image everything. See Valm et al https://www.ncbi.nlm.nih.gov/pubmed/27391327 for one example of high multiplexing (not tyramides) using all laser lines and detectors on a spectral confocal microscope.

I see in SuperBoost PDF only 7 colors (7 Alexa's)

https://tools.thermofisher.com/content/sfs/manuals/tyramide_superboost_kits_man.pdf

to get to 10 (ore more) could use biotin. A 1998 paper had do-it-yourself tyramide hapten synthesis
http://jhc.sagepub.com/content/46/6/771.full

Sure, could also do this single color, either add the signals, or photobleach. This could be useful on the MilliporeSigma/EMD Millipore/CellASIC ONIX microperfusion platform (since ONIX only has a few reservoirs and they seem to lack interest in making plates and manifolds for use with 2 or more controllers).

Another option for quantitation: wait for Garry P. Nolan (Stanford Univ.) to publish his immuofluorescence alternative to MIBI-ToF, or collaborate with him to use his MIBI-ToF or buy a Fluidigm imaging CyTOF.

enjoy,

George


On 10/13/2016 11:52 AM, Claire Brown, Dr. wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

I wonder if anyone knows if the new ThermoFisher Tyramide SuperBoost kits are quantitative?

 

https://www.thermofisher.com/order/catalog/product/B40915

 

 

I would assume that keeping the reaction conditions identical from sample to sample would be really important in stopping the HRP reaction after a fixed time but with such amplification is there any way relative signals could be quantified?

 

I would love to hear what people think.

 

Sincerely,

 

Claire

 

 


-- 


George McNamara, PhD
Houston, TX 77054
[hidden email]
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/
http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650

-- 


George McNamara, PhD
Houston, TX 77054
[hidden email]
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/
http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650

Mark Cannell-2 Mark Cannell-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Tyramide superboost kits quantitative?

In reply to this post by Sripad Ram-2
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi

This contains some references that may be relevant to the question of how far radicals move?

doi: 10.1177/31.7.6304184J Histochem Cytochemvol. 31 no. 7 945-951

HTH Mark
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Hi George,
As always super informative. Is it just me or do others also think that its high time someone invents a device download all that info that's sitting in your head...?

Very interesting comment about the diffusion radius of the activated TSA radical being less than the diffraction limit. I think that is a very important parameter. Just wondering if you know a reference for that? 

Perkin Elmer recently showed some correlative fluorescence and EM data which suggests similar range of specificity for their TSA staining. 


Thanks!

Sripad Ram





On Thu, Oct 13, 2016 at 11:24 AM, George McNamara <[hidden email]> wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Claire,

Depending on temperature, constant reaction time may be bad. Minimal issue is always run on a warm plate (that is perfectly uniform temperature ... there are slide thermometers for this), incubators or ovens.

I suspect the TSA reaction runs to completion usually, with one of these running out:

* active HRP enzyme molecules (self destructs with some probability for every H2O2 catalytic cycle)

* tyrosine "docking sites" ... both in cells/tissues, the antibody molecules, and the HRPs (an advantage for poly HRP)

* fluorescent tyramide (least likely?)

* H2O2 (also unlikely?)


I believe PerkinElmer's Opal kits for multiplexing deliberately strips the detection Ab-polyHRP, before the next cycle, but have not been able to get data from PerkinElmer on this (and did not need TSA while I worked in a research lab here in Houston).


Many IHC reagent companies sell HRP polymers conjugated to streptavidin or secondary antibodies. For example
STREPTAVIDIN POLY-HRP80 CONJUGATE (plus stabilizer diluent)
https://www.fitzgerald-fii.com/streptavidin-poly-hrp80-conjugate-65r-s118.html

Yes, you can replace components in the kit with other reagents ... and conversely, use components in the kit for other purposes. An interesting "head to head" is fluorescent tyramide vs DAB IHC, or whatever favorite HRP substrate detection reagent your friendly neighborhood pathologists like.

I'm a big fan of TSA ever  since Phil Moen of NEN (acquired by PerkinElmer) spoke about and demo'd it at a Cold Spring Harbor fluorescence in situ/immuno courses. Phil mentioned the diffusion radius of the activated tyramide radical is less than the diffraction limit of confocal microscopy (i.e. 100 nm radius, 200 nm diameter).

One potential limitation on quantitation: zero reactable tyrosines implies zero signal ... as I mentioned above, both the Ab and HRPs have tyrosines (some even on the surface if I recall correctly).

Big benefit: when done right, can dilute the primary antibody 100 fold (maybe more ... try it), saving money. In fact, failing to do so will result in massive background.

Automation: in principle, automating TSA should improve reproducibility (assuming constant temperature for the reaction, and either constant enzyme reaction time or "run to completion"). However, not paying attention to the cost of the reagents (or stability or lack thereof) could result in one well, $100 in reagents --> one slide).

One of my U Miami customers published a super-duper brightness CD4 and CD8 2 color T-cells (also CD4 and FoxP3) paper ... they killed off the HRP molecules (and before that, endogenous peroxidases) with a terrific product (both great name and works well): PeroxAbolish   http://biocare.net/product/peroxabolish/

https://www.ncbi.nlm.nih.gov/pubmed/21929847

Cell Transplant. 2012;21(1):113-25. doi: 10.3727/096368911X586747. Epub 2011 Sep 
16.

Quantitative in situ analysis of FoxP3+ T regulatory cells on transplant tissue
using laser scanning cytometry.

Takahashi H(1), Ruiz P, Ricordi C, Delacruz V, Miki A, Mita A, Misawa R, Barker
S, Burke GW, Tzakis AG, Ichii H.

Author information: 
(1)Miami Transplant Institute, University of Miami Leonard M. Miller School of
Medicine, Miami, FL, USA.

There is abundant evidence that immune cells infiltrating into a transplanted
organ play a critical role for destructive inflammatory or regulatory immune
reactions. Quantitative in situ analysis (i.e., in tissue sections) of immune
cells remains challenging due to a lack of objective methodology. Laser scanning 
cytometry (LSC) is an imaging-based methodology that performs quantitative
measurements on fluorescently and/ or chromatically stained tissue or cellular
specimens at a single-cell level. In this study, we have developed a novel
objective method for analysis of immune cells, including Foxp3(+) T regulatory
cells (Tregs), on formalin-fixed /paraffin-embedded (FFPE) transplant biopsy
sections using iCys® Research Imaging Cytometer. The development of multiple
immunofluorescent staining was established using FFPE human tonsil sample. The
CD4/CD8 ratio and the population of Tregs among CD4(+) cells were analyzed using 
iCys and compared with the results from conventional flow cytometry analysis
(FCM). Our multiple immunofluorescent staining techniques allow obtaining clear
staining on FFPE sections. The CD4/CD8 ratio analyzed by iCys was concordant with
those obtained by FCM. This method was also applicable for liver, small
intestine, kidney, pancreas, and heart transplant biopsy sections and provide an 
objective quantification of Tregs within the grafts.

DOI: 10.3727/096368911X586747 
PMCID: PMC3777543
PMID: 21929847  [PubMed - indexed for MEDLINE]
and my 2010 book chapter (Yuste 2010 CSHL Press chapter 15 - Imaging in Neurosciences) has a protocol (I would use Mol Probes/ThermoFisher or PerkinElmers ... can probably crosslink the Ab-HRPpolymer onto the cells/tissue instead of their strip).

I also want to acknowledge Molecular Probes - specifically Mike Janes and his team - for visiting Miami (several years ago) for a workshop featuring TSA and Bacmam2.0. Hopefully ThermoFisher still lets the Mol Probes folks out of the lab to do similar workshops.

Speculation (spectral-ation - I currently do not work in a lab, so not going to test this anytime soon -- maybe someone from Mol Probes can try it):

Single molecule localization staining ... or "rainbow single antigen counting by TSA" idea (this will be tedious to do by hand! Payoff: count everything):
=>I suggest using 35 mm imaging dish, so that wash steps can be very large volume (2 mL), and expensive reagent steps just the imaging area (7, 10, 14, 20, 28 mm diameter as appropriate for your expt).
1. incubate primary antibody at ~1/10th of saturation.
2. wash (extensively).
3. incubate secondary antibody-polyHRP.
4. wash extensively.
5. detect with Tyramide color #1.
6. kill HRP, i.e. with PeroxAbolish (http://biocare.net/product/peroxabolish/) AND I suggest gently crosslinking the 2ndAb-HRP-polymer into the specimen (see also Expansion Microscopy approaches).
7. OK, may want to image at this point, ideally with a way to refind the same place later.
*** Repeat 1-6 with different color, optionally also #7.
*. Image everything. See Valm et al https://www.ncbi.nlm.nih.gov/pubmed/27391327 for one example of high multiplexing (not tyramides) using all laser lines and detectors on a spectral confocal microscope.

I see in SuperBoost PDF only 7 colors (7 Alexa's)

https://tools.thermofisher.com/content/sfs/manuals/tyramide_superboost_kits_man.pdf

to get to 10 (ore more) could use biotin. A 1998 paper had do-it-yourself tyramide hapten synthesis
http://jhc.sagepub.com/content/46/6/771.full

Sure, could also do this single color, either add the signals, or photobleach. This could be useful on the MilliporeSigma/EMD Millipore/CellASIC ONIX microperfusion platform (since ONIX only has a few reservoirs and they seem to lack interest in making plates and manifolds for use with 2 or more controllers).

Another option for quantitation: wait for Garry P. Nolan (Stanford Univ.) to publish his immuofluorescence alternative to MIBI-ToF, or collaborate with him to use his MIBI-ToF or buy a Fluidigm imaging CyTOF.

enjoy,

George


On 10/13/2016 11:52 AM, Claire Brown, Dr. wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

I wonder if anyone knows if the new ThermoFisher Tyramide SuperBoost kits are quantitative?

 

https://www.thermofisher.com/order/catalog/product/B40915

 

 

I would assume that keeping the reaction conditions identical from sample to sample would be really important in stopping the HRP reaction after a fixed time but with such amplification is there any way relative signals could be quantified?

 

I would love to hear what people think.

 

Sincerely,

 

Claire

 

 


-- 


George McNamara, PhD
Houston, TX 77054
[hidden email]
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/
http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650



Mark  B. Cannell Ph.D. FRSNZ FISHR
Professor of Cardiac Cell Biology
School of Physiology &  Pharmacology
Faculty of Biomedical Sciences
University of Bristol
Bristol
BS8 1TD UK




George McNamara George McNamara
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Tyramide superboost kits quantitative?

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Mark,

thanks for the find! the kidney GBM laminae rarae's ~60 nm periodicity could make this a nice nanoscopic calibration standard.

I also realized -- from their mention of 2 minute incubation times to limit the DAB + H2O2 reaction -- that my "Single molecule localization staining ... or "rainbow single antigen counting by TSA" idea below could be greatly simplified: detect with tyramide color 1 briefly, then color 2, then color 3, etc.

George


On 10/14/2016 3:07 AM, Mark Cannell wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi

This contains some references that may be relevant to the question of how far radicals move?

doi: 10.1177/31.7.6304184J Histochem Cytochemvol. 31 no. 7 945-951

HTH Mark
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Hi George,
As always super informative. Is it just me or do others also think that its high time someone invents a device download all that info that's sitting in your head...?

Very interesting comment about the diffusion radius of the activated TSA radical being less than the diffraction limit. I think that is a very important parameter. Just wondering if you know a reference for that? 

Perkin Elmer recently showed some correlative fluorescence and EM data which suggests similar range of specificity for their TSA staining. 


Thanks!

Sripad Ram





On Thu, Oct 13, 2016 at 11:24 AM, George McNamara <[hidden email]> wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Claire,

Depending on temperature, constant reaction time may be bad. Minimal issue is always run on a warm plate (that is perfectly uniform temperature ... there are slide thermometers for this), incubators or ovens.

I suspect the TSA reaction runs to completion usually, with one of these running out:

* active HRP enzyme molecules (self destructs with some probability for every H2O2 catalytic cycle)

* tyrosine "docking sites" ... both in cells/tissues, the antibody molecules, and the HRPs (an advantage for poly HRP)

* fluorescent tyramide (least likely?)

* H2O2 (also unlikely?)


I believe PerkinElmer's Opal kits for multiplexing deliberately strips the detection Ab-polyHRP, before the next cycle, but have not been able to get data from PerkinElmer on this (and did not need TSA while I worked in a research lab here in Houston).


Many IHC reagent companies sell HRP polymers conjugated to streptavidin or secondary antibodies. For example
STREPTAVIDIN POLY-HRP80 CONJUGATE (plus stabilizer diluent)
https://www.fitzgerald-fii.com/streptavidin-poly-hrp80-conjugate-65r-s118.html

Yes, you can replace components in the kit with other reagents ... and conversely, use components in the kit for other purposes. An interesting "head to head" is fluorescent tyramide vs DAB IHC, or whatever favorite HRP substrate detection reagent your friendly neighborhood pathologists like.

I'm a big fan of TSA ever  since Phil Moen of NEN (acquired by PerkinElmer) spoke about and demo'd it at a Cold Spring Harbor fluorescence in situ/immuno courses. Phil mentioned the diffusion radius of the activated tyramide radical is less than the diffraction limit of confocal microscopy (i.e. 100 nm radius, 200 nm diameter).

One potential limitation on quantitation: zero reactable tyrosines implies zero signal ... as I mentioned above, both the Ab and HRPs have tyrosines (some even on the surface if I recall correctly).

Big benefit: when done right, can dilute the primary antibody 100 fold (maybe more ... try it), saving money. In fact, failing to do so will result in massive background.

Automation: in principle, automating TSA should improve reproducibility (assuming constant temperature for the reaction, and either constant enzyme reaction time or "run to completion"). However, not paying attention to the cost of the reagents (or stability or lack thereof) could result in one well, $100 in reagents --> one slide).

One of my U Miami customers published a super-duper brightness CD4 and CD8 2 color T-cells (also CD4 and FoxP3) paper ... they killed off the HRP molecules (and before that, endogenous peroxidases) with a terrific product (both great name and works well): PeroxAbolish   http://biocare.net/product/peroxabolish/

https://www.ncbi.nlm.nih.gov/pubmed/21929847

Cell Transplant. 2012;21(1):113-25. doi: 10.3727/096368911X586747. Epub 2011 Sep 
16.

Quantitative in situ analysis of FoxP3+ T regulatory cells on transplant tissue
using laser scanning cytometry.

Takahashi H(1), Ruiz P, Ricordi C, Delacruz V, Miki A, Mita A, Misawa R, Barker
S, Burke GW, Tzakis AG, Ichii H.

Author information: 
(1)Miami Transplant Institute, University of Miami Leonard M. Miller School of
Medicine, Miami, FL, USA.

There is abundant evidence that immune cells infiltrating into a transplanted
organ play a critical role for destructive inflammatory or regulatory immune
reactions. Quantitative in situ analysis (i.e., in tissue sections) of immune
cells remains challenging due to a lack of objective methodology. Laser scanning 
cytometry (LSC) is an imaging-based methodology that performs quantitative
measurements on fluorescently and/ or chromatically stained tissue or cellular
specimens at a single-cell level. In this study, we have developed a novel
objective method for analysis of immune cells, including Foxp3(+) T regulatory
cells (Tregs), on formalin-fixed /paraffin-embedded (FFPE) transplant biopsy
sections using iCys® Research Imaging Cytometer. The development of multiple
immunofluorescent staining was established using FFPE human tonsil sample. The
CD4/CD8 ratio and the population of Tregs among CD4(+) cells were analyzed using 
iCys and compared with the results from conventional flow cytometry analysis
(FCM). Our multiple immunofluorescent staining techniques allow obtaining clear
staining on FFPE sections. The CD4/CD8 ratio analyzed by iCys was concordant with
those obtained by FCM. This method was also applicable for liver, small
intestine, kidney, pancreas, and heart transplant biopsy sections and provide an 
objective quantification of Tregs within the grafts.

DOI: 10.3727/096368911X586747 
PMCID: PMC3777543
PMID: 21929847  [PubMed - indexed for MEDLINE]
and my 2010 book chapter (Yuste 2010 CSHL Press chapter 15 - Imaging in Neurosciences) has a protocol (I would use Mol Probes/ThermoFisher or PerkinElmers ... can probably crosslink the Ab-HRPpolymer onto the cells/tissue instead of their strip).

I also want to acknowledge Molecular Probes - specifically Mike Janes and his team - for visiting Miami (several years ago) for a workshop featuring TSA and Bacmam2.0. Hopefully ThermoFisher still lets the Mol Probes folks out of the lab to do similar workshops.

Speculation (spectral-ation - I currently do not work in a lab, so not going to test this anytime soon -- maybe someone from Mol Probes can try it):

Single molecule localization staining ... or "rainbow single antigen counting by TSA" idea (this will be tedious to do by hand! Payoff: count everything):
=>I suggest using 35 mm imaging dish, so that wash steps can be very large volume (2 mL), and expensive reagent steps just the imaging area (7, 10, 14, 20, 28 mm diameter as appropriate for your expt).
1. incubate primary antibody at ~1/10th of saturation.
2. wash (extensively).
3. incubate secondary antibody-polyHRP.
4. wash extensively.
5. detect with Tyramide color #1.
6. kill HRP, i.e. with PeroxAbolish (http://biocare.net/product/peroxabolish/) AND I suggest gently crosslinking the 2ndAb-HRP-polymer into the specimen (see also Expansion Microscopy approaches).
7. OK, may want to image at this point, ideally with a way to refind the same place later.
*** Repeat 1-6 with different color, optionally also #7.
*. Image everything. See Valm et al https://www.ncbi.nlm.nih.gov/pubmed/27391327 for one example of high multiplexing (not tyramides) using all laser lines and detectors on a spectral confocal microscope.

I see in SuperBoost PDF only 7 colors (7 Alexa's)

https://tools.thermofisher.com/content/sfs/manuals/tyramide_superboost_kits_man.pdf

to get to 10 (ore more) could use biotin. A 1998 paper had do-it-yourself tyramide hapten synthesis
http://jhc.sagepub.com/content/46/6/771.full

Sure, could also do this single color, either add the signals, or photobleach. This could be useful on the MilliporeSigma/EMD Millipore/CellASIC ONIX microperfusion platform (since ONIX only has a few reservoirs and they seem to lack interest in making plates and manifolds for use with 2 or more controllers).

Another option for quantitation: wait for Garry P. Nolan (Stanford Univ.) to publish his immuofluorescence alternative to MIBI-ToF, or collaborate with him to use his MIBI-ToF or buy a Fluidigm imaging CyTOF.

enjoy,

George


On 10/13/2016 11:52 AM, Claire Brown, Dr. wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

I wonder if anyone knows if the new ThermoFisher Tyramide SuperBoost kits are quantitative?

 

https://www.thermofisher.com/order/catalog/product/B40915

 

 

I would assume that keeping the reaction conditions identical from sample to sample would be really important in stopping the HRP reaction after a fixed time but with such amplification is there any way relative signals could be quantified?

 

I would love to hear what people think.

 

Sincerely,

 

Claire

 

 


-- 


George McNamara, PhD
Houston, TX 77054
[hidden email]
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/
http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650

Mark  B. Cannell Ph.D. FRSNZ FISHR Professor of Cardiac Cell Biology School of Physiology &  Pharmacology Faculty of Biomedical Sciences University of Bristol Bristol BS8 1TD UK
-- 


George McNamara, PhD
Houston, TX 77054
[hidden email]
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/
http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650

Sripad Ram-2 Sripad Ram-2
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: Tyramide superboost kits quantitative?

***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Mark and George,
Thanks for the feedback and comments! 

Sripad


On Fri, Oct 14, 2016 at 5:40 AM, George McNamara <[hidden email]> wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Mark,

thanks for the find! the kidney GBM laminae rarae's ~60 nm periodicity could make this a nice nanoscopic calibration standard.

I also realized -- from their mention of 2 minute incubation times to limit the DAB + H2O2 reaction -- that my "Single molecule localization staining ... or "rainbow single antigen counting by TSA" idea below could be greatly simplified: detect with tyramide color 1 briefly, then color 2, then color 3, etc.

George


On 10/14/2016 3:07 AM, Mark Cannell wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi

This contains some references that may be relevant to the question of how far radicals move?

doi: 10.1177/31.7.6304184J Histochem Cytochem July 1983 vol. 31 no. 7 945-951

HTH Mark
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****
Hi George,
As always super informative. Is it just me or do others also think that its high time someone invents a device download all that info that's sitting in your head...?

Very interesting comment about the diffusion radius of the activated TSA radical being less than the diffraction limit. I think that is a very important parameter. Just wondering if you know a reference for that? 

Perkin Elmer recently showed some correlative fluorescence and EM data which suggests similar range of specificity for their TSA staining. 


Thanks!

Sripad Ram





On Thu, Oct 13, 2016 at 11:24 AM, George McNamara <[hidden email]> wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

Hi Claire,

Depending on temperature, constant reaction time may be bad. Minimal issue is always run on a warm plate (that is perfectly uniform temperature ... there are slide thermometers for this), incubators or ovens.

I suspect the TSA reaction runs to completion usually, with one of these running out:

* active HRP enzyme molecules (self destructs with some probability for every H2O2 catalytic cycle)

* tyrosine "docking sites" ... both in cells/tissues, the antibody molecules, and the HRPs (an advantage for poly HRP)

* fluorescent tyramide (least likely?)

* H2O2 (also unlikely?)


I believe PerkinElmer's Opal kits for multiplexing deliberately strips the detection Ab-polyHRP, before the next cycle, but have not been able to get data from PerkinElmer on this (and did not need TSA while I worked in a research lab here in Houston).


Many IHC reagent companies sell HRP polymers conjugated to streptavidin or secondary antibodies. For example
STREPTAVIDIN POLY-HRP80 CONJUGATE (plus stabilizer diluent)
https://www.fitzgerald-fii.com/streptavidin-poly-hrp80-conjugate-65r-s118.html

Yes, you can replace components in the kit with other reagents ... and conversely, use components in the kit for other purposes. An interesting "head to head" is fluorescent tyramide vs DAB IHC, or whatever favorite HRP substrate detection reagent your friendly neighborhood pathologists like.

I'm a big fan of TSA ever  since Phil Moen of NEN (acquired by PerkinElmer) spoke about and demo'd it at a Cold Spring Harbor fluorescence in situ/immuno courses. Phil mentioned the diffusion radius of the activated tyramide radical is less than the diffraction limit of confocal microscopy (i.e. 100 nm radius, 200 nm diameter).

One potential limitation on quantitation: zero reactable tyrosines implies zero signal ... as I mentioned above, both the Ab and HRPs have tyrosines (some even on the surface if I recall correctly).

Big benefit: when done right, can dilute the primary antibody 100 fold (maybe more ... try it), saving money. In fact, failing to do so will result in massive background.

Automation: in principle, automating TSA should improve reproducibility (assuming constant temperature for the reaction, and either constant enzyme reaction time or "run to completion"). However, not paying attention to the cost of the reagents (or stability or lack thereof) could result in one well, $100 in reagents --> one slide).

One of my U Miami customers published a super-duper brightness CD4 and CD8 2 color T-cells (also CD4 and FoxP3) paper ... they killed off the HRP molecules (and before that, endogenous peroxidases) with a terrific product (both great name and works well): PeroxAbolish   http://biocare.net/product/peroxabolish/

https://www.ncbi.nlm.nih.gov/pubmed/21929847

Cell Transplant. 2012;21(1):113-25. doi: 10.3727/096368911X586747. Epub 2011 Sep 
16.

Quantitative in situ analysis of FoxP3+ T regulatory cells on transplant tissue
using laser scanning cytometry.

Takahashi H(1), Ruiz P, Ricordi C, Delacruz V, Miki A, Mita A, Misawa R, Barker
S, Burke GW, Tzakis AG, Ichii H.

Author information: 
(1)Miami Transplant Institute, University of Miami Leonard M. Miller School of
Medicine, Miami, FL, USA.

There is abundant evidence that immune cells infiltrating into a transplanted
organ play a critical role for destructive inflammatory or regulatory immune
reactions. Quantitative in situ analysis (i.e., in tissue sections) of immune
cells remains challenging due to a lack of objective methodology. Laser scanning 
cytometry (LSC) is an imaging-based methodology that performs quantitative
measurements on fluorescently and/ or chromatically stained tissue or cellular
specimens at a single-cell level. In this study, we have developed a novel
objective method for analysis of immune cells, including Foxp3(+) T regulatory
cells (Tregs), on formalin-fixed /paraffin-embedded (FFPE) transplant biopsy
sections using iCys® Research Imaging Cytometer. The development of multiple
immunofluorescent staining was established using FFPE human tonsil sample. The
CD4/CD8 ratio and the population of Tregs among CD4(+) cells were analyzed using 
iCys and compared with the results from conventional flow cytometry analysis
(FCM). Our multiple immunofluorescent staining techniques allow obtaining clear
staining on FFPE sections. The CD4/CD8 ratio analyzed by iCys was concordant with
those obtained by FCM. This method was also applicable for liver, small
intestine, kidney, pancreas, and heart transplant biopsy sections and provide an 
objective quantification of Tregs within the grafts.

DOI: 10.3727/096368911X586747 
PMCID: PMC3777543
PMID: 21929847  [PubMed - indexed for MEDLINE]
and my 2010 book chapter (Yuste 2010 CSHL Press chapter 15 - Imaging in Neurosciences) has a protocol (I would use Mol Probes/ThermoFisher or PerkinElmers ... can probably crosslink the Ab-HRPpolymer onto the cells/tissue instead of their strip).

I also want to acknowledge Molecular Probes - specifically Mike Janes and his team - for visiting Miami (several years ago) for a workshop featuring TSA and Bacmam2.0. Hopefully ThermoFisher still lets the Mol Probes folks out of the lab to do similar workshops.

Speculation (spectral-ation - I currently do not work in a lab, so not going to test this anytime soon -- maybe someone from Mol Probes can try it):

Single molecule localization staining ... or "rainbow single antigen counting by TSA" idea (this will be tedious to do by hand! Payoff: count everything):
=>I suggest using 35 mm imaging dish, so that wash steps can be very large volume (2 mL), and expensive reagent steps just the imaging area (7, 10, 14, 20, 28 mm diameter as appropriate for your expt).
1. incubate primary antibody at ~1/10th of saturation.
2. wash (extensively).
3. incubate secondary antibody-polyHRP.
4. wash extensively.
5. detect with Tyramide color #1.
6. kill HRP, i.e. with PeroxAbolish (http://biocare.net/product/peroxabolish/) AND I suggest gently crosslinking the 2ndAb-HRP-polymer into the specimen (see also Expansion Microscopy approaches).
7. OK, may want to image at this point, ideally with a way to refind the same place later.
*** Repeat 1-6 with different color, optionally also #7.
*. Image everything. See Valm et al https://www.ncbi.nlm.nih.gov/pubmed/27391327 for one example of high multiplexing (not tyramides) using all laser lines and detectors on a spectral confocal microscope.

I see in SuperBoost PDF only 7 colors (7 Alexa's)

https://tools.thermofisher.com/content/sfs/manuals/tyramide_superboost_kits_man.pdf

to get to 10 (ore more) could use biotin. A 1998 paper had do-it-yourself tyramide hapten synthesis
http://jhc.sagepub.com/content/46/6/771.full

Sure, could also do this single color, either add the signals, or photobleach. This could be useful on the MilliporeSigma/EMD Millipore/CellASIC ONIX microperfusion platform (since ONIX only has a few reservoirs and they seem to lack interest in making plates and manifolds for use with 2 or more controllers).

Another option for quantitation: wait for Garry P. Nolan (Stanford Univ.) to publish his immuofluorescence alternative to MIBI-ToF, or collaborate with him to use his MIBI-ToF or buy a Fluidigm imaging CyTOF.

enjoy,

George


On 10/13/2016 11:52 AM, Claire Brown, Dr. wrote:
***** To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. *****

I wonder if anyone knows if the new ThermoFisher Tyramide SuperBoost kits are quantitative?

 

https://www.thermofisher.com/order/catalog/product/B40915

 

 

I would assume that keeping the reaction conditions identical from sample to sample would be really important in stopping the HRP reaction after a fixed time but with such amplification is there any way relative signals could be quantified?

 

I would love to hear what people think.

 

Sincerely,

 

Claire

 

 


-- 


George McNamara, PhD
Houston, TX 77054
[hidden email]
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/
http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650

Mark  B. Cannell Ph.D. FRSNZ FISHR Professor of Cardiac Cell Biology School of Physiology &  Pharmacology Faculty of Biomedical Sciences University of Bristol Bristol BS8 1TD UK
-- 


George McNamara, PhD
Houston, TX 77054
[hidden email]
https://www.linkedin.com/in/georgemcnamara
https://works.bepress.com/gmcnamara/75/
http://www.ncbi.nlm.nih.gov/myncbi/browse/collection/44962650