UPS for Zeiss 510 Meta

classic Classic list List threaded Threaded
♦
♦
8 messages Options
yuansheng sun yuansheng sun
Reply | Threaded
Open this post in threaded view
♦
♦
|

UPS for Zeiss 510 Meta

Dear list members,

We have a Zeiss 510 Meta system with the Argon, DPSS 561 and HeNe 633
lasers.  We like to buy a UPS for the system to protect the lasers
from power failures.  Any suggestions?
Many thanks in advance.

Best regards,
sheng
Armstrong, Brian Armstrong, Brian
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: UPS for Zeiss 510 Meta

Hello, I believe we use a PowerWare 9170. I think they are stackable so
you can buy what you want/need.
http://www.42u.com/powerware-9170-ups.htm


Brian Armstrong PhD
Manager, Light Microscopy and Digital Imaging Core
Beckman Research Institute of the City of Hope
Neuroscience
X62872
-----Original Message-----
From: Confocal Microscopy List [mailto:[hidden email]]
On Behalf Of yuansheng sun
Sent: Thursday, June 03, 2010 2:59 PM
To: [hidden email]
Subject: UPS for Zeiss 510 Meta

Dear list members,

We have a Zeiss 510 Meta system with the Argon, DPSS 561 and HeNe 633
lasers.  We like to buy a UPS for the system to protect the lasers
from power failures.  Any suggestions?
Many thanks in advance.

Best regards,
sheng


---------------------------------------------------------------------
SECURITY/CONFIDENTIALITY WARNING:  
This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender.

---------------------------------------------------------------------
Mark A Townley Mark A Townley
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: UPS for Zeiss 510 Meta

In reply to this post by yuansheng sun
We have the same system and have had a GE (LP Series) UPS for it for  
the last 3 years. No problems so far.

Mark


* yuansheng sun <[hidden email]> [Thu 03 Jun 2010 05:59:13 PM EDT]:

> Dear list members,
>
> We have a Zeiss 510 Meta system with the Argon, DPSS 561 and HeNe 633
> lasers.  We like to buy a UPS for the system to protect the lasers
> from power failures.  Any suggestions?
> Many thanks in advance.
>
> Best regards,
> sheng
>
G. Esteban Fernandez G. Esteban Fernandez
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: UPS for Zeiss 510 Meta

In reply to this post by yuansheng sun
U.S. Tech sells one specifically designed for Zeiss confocals.
Contact Pam Ford at [hidden email], 800-783-8187

(no commercial interest)

-Esteban


On Thu, Jun 3, 2010 at 2:59 PM, yuansheng sun <[hidden email]> wrote:

> Dear list members,
>
> We have a Zeiss 510 Meta system with the Argon, DPSS 561 and HeNe 633
> lasers.  We like to buy a UPS for the system to protect the lasers
> from power failures.  Any suggestions?
> Many thanks in advance.
>
> Best regards,
> sheng
>
Boswell, Carl A - (cboswell) Boswell, Carl A - (cboswell)
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: UPS for Zeiss 510 Meta

In reply to this post by yuansheng sun
Hi Sheng,

We have a Powerware PW9125 that has performed flawlessly for 5+ years.  We
actually had a power outage in the middle of an imaging system and the UPS
provided plenty of time to calmly shut the system down.  We got it from
Power Pros: www.powerprosinc.com.

At the least, be sure to do some comparison shopping as the UPS recommended
by Zeiss was maybe 50% more than what we finally paid.  There is nothing
special about a UPS "specifically" designed for the system.  The specs are
straighforward and the the rep at Power Pros had no problem configuring what
we needed.

All imaging systems like this should have a roubust UPS between it and house
line voltage.    One of the benefits of a UPS is its ability to act a as a
line polisher, maintaining a very steady electrical input to the instrument.
Aside from blatant power outages, surges and drops in line voltage can raise
hell with software and hardware function and longevity.

Good luck,
c

Carl A. Boswell, Ph.D.
Molecular and Cellular Biology
University of Arizona
520-954-7053
FAX 520-621-3709
----- Original Message -----
From: "yuansheng sun" <[hidden email]>
To: <[hidden email]>
Sent: Thursday, June 03, 2010 2:59 PM
Subject: UPS for Zeiss 510 Meta


> Dear list members,
>
> We have a Zeiss 510 Meta system with the Argon, DPSS 561 and HeNe 633
> lasers.  We like to buy a UPS for the system to protect the lasers
> from power failures.  Any suggestions?
> Many thanks in advance.
>
> Best regards,
> sheng
>
RICHARD BURRY RICHARD BURRY
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: UPS for Zeiss 510 Meta

Sheng
 
I second Carl.  We have UPS systems for all of our confocals and the best function is polishing.  The microscope actually runs off the battery and the unit keeps the battery charged.  The result (no experimental proof) is longer life on lasers and computers.
 
Dick Burry

----- Original Message -----
From: Carl Boswell <[hidden email]>
Date: Thursday, June 3, 2010 8:18 pm
Subject: Re: UPS for Zeiss 510 Meta
To: [hidden email]

> Hi Sheng,
>
> We have a Powerware PW9125 that has performed flawlessly for 5+
> years.  We
> actually had a power outage in the middle of an imaging system
> and the UPS
> provided plenty of time to calmly shut the system down.  We
> got it from
> Power Pros: www.powerprosinc.com.
>
> At the least, be sure to do some comparison shopping as the UPS
> recommended
> by Zeiss was maybe 50% more than what we finally paid. 
> There is nothing
> special about a UPS "specifically" designed for the
> system.  The specs are
> straighforward and the the rep at Power Pros had no problem
> configuring what
> we needed.
>
> All imaging systems like this should have a roubust UPS between
> it and house
> line voltage.    One of the benefits of a UPS is
> its ability to act a as a
> line polisher, maintaining a very steady electrical input to the
> instrument.
> Aside from blatant power outages, surges and drops in line
> voltage can raise
> hell with software and hardware function and longevity.
>
> Good luck,
> c
>
> Carl A. Boswell, Ph.D.
> Molecular and Cellular Biology
> University of Arizona
> 520-954-7053
> FAX 520-621-3709
> ----- Original Message -----
> From: "yuansheng sun" <[hidden email]>
> To: <[hidden email]>
> Sent: Thursday, June 03, 2010 2:59 PM
> Subject: UPS for Zeiss 510 Meta
>
>
> > Dear list members,
> >
> > We have a Zeiss 510 Meta system with the Argon, DPSS 561 and
> HeNe 633
> > lasers.  We like to buy a UPS for the system to protect
> the lasers
> > from power failures.  Any suggestions?
> > Many thanks in advance.
> >
> > Best regards,
> > sheng
> >
>
>
> --
> BEGIN-ANTISPAM-VOTING-LINKS
> ------------------------------------------------------
>
> Teach CanIt if this mail (ID 1047546538) is spam:
> Spam:       
> https://antispam.osu.edu/b.php?i=1047546538&m=7800f87eaac5&c=sNot spam:    https://antispam.osu.edu/b.php?i=1047546538&m=7800f87eaac5&c=n
> Forget vote:
> https://antispam.osu.edu/b.php?i=1047546538&m=7800f87eaac5&c=f---
> ---------------------------------------------------
> END-ANTISPAM-VOTING-LINKS
>
Mark A Townley Mark A Townley
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: UPS for Zeiss 510 Meta

In reply to this post by G. Esteban Fernandez
The GE 51002Y (LP Series) UPS I mentioned earlier was purchased  
through Pam Ford at U.S. Tech.

Mark


* "G. Esteban Fernandez" <[hidden email]> [Thu 03 Jun  
2010 06:36:23 PM EDT]:

> U.S. Tech sells one specifically designed for Zeiss confocals.
> Contact Pam Ford at [hidden email], 800-783-8187
>
> (no commercial interest)
>
> -Esteban
>
>
> On Thu, Jun 3, 2010 at 2:59 PM, yuansheng sun  
> <[hidden email]> wrote:
>> Dear list members,
>>
>> We have a Zeiss 510 Meta system with the Argon, DPSS 561 and HeNe 633
>> lasers.  We like to buy a UPS for the system to protect the lasers
>> from power failures.  Any suggestions?
>> Many thanks in advance.
>>
>> Best regards,
>> sheng
>>
>
yuansheng sun yuansheng sun
Reply | Threaded
Open this post in threaded view
♦
♦
|

Re: UPS for Zeiss 510 Meta

Many thanks for all the inputs, very helpful.  - sheng


On Fri, Jun 4, 2010 at 10:01 AM, Mark A Townley
<[hidden email]> wrote:

> The GE 51002Y (LP Series) UPS I mentioned earlier was purchased through Pam
> Ford at U.S. Tech.
>
> Mark
>
>
> * "G. Esteban Fernandez" <[hidden email]> [Thu 03 Jun 2010
> 06:36:23 PM EDT]:
>
>> U.S. Tech sells one specifically designed for Zeiss confocals.
>> Contact Pam Ford at [hidden email], 800-783-8187
>>
>> (no commercial interest)
>>
>> -Esteban
>>
>>
>> On Thu, Jun 3, 2010 at 2:59 PM, yuansheng sun <[hidden email]>
>> wrote:
>>>
>>> Dear list members,
>>>
>>> We have a Zeiss 510 Meta system with the Argon, DPSS 561 and HeNe 633
>>> lasers.  We like to buy a UPS for the system to protect the lasers
>>> from power failures.  Any suggestions?
>>> Many thanks in advance.
>>>
>>> Best regards,
>>> sheng
>>>
>>
>