Michael Giacomelli |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I recently discovered an old laser in storage. Unfortunately the previous users had not drained the coolant lines, resulting in several years corrosion blocking the lines. I flushed them out as best I could and found a grey metal slurry that was fairly thick in place of what was once the coolant. Unfortunately, even with flushing I can't flow anything through the system using a normal thermotek chiller. Are there any options in this situation to unclog a laser? Looking at the material in the line, its fairly diffuse and does not appear all that solid. Is it possible to use something like a weak acid to dissolve any solid material blocking the line? Or some other solvent/cleaner? Mike |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I've used household vinegar for this, or you can make up 5% acetic acid from lab stock. If you can get any flow at all, just keep running solution through it. If you're lucky, the vinegar will dissolve out any mineral deposits and also help loosen up any biological gunk. Once you have things flowing well, flush with regular water. Finally, put Optishield or similar into your final coolant water to help keep this from happening in the future. I'm sure others will chime in with their own magic recipes for loosening up stuff like this, so give it all a try and you should hit on something that works. Best of luck! Craig Brideau On Mon, Nov 23, 2015 at 2:46 PM, Michael Giacomelli <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I recently discovered an old laser in storage. Unfortunately the > previous users had not drained the coolant lines, resulting in several > years corrosion blocking the lines. I flushed them out as best I > could and found a grey metal slurry that was fairly thick in place of > what was once the coolant. Unfortunately, even with flushing I can't > flow anything through the system using a normal thermotek chiller. > > Are there any options in this situation to unclog a laser? Looking at > the material in the line, its fairly diffuse and does not appear all > that solid. Is it possible to use something like a weak acid to > dissolve any solid material blocking the line? Or some other > solvent/cleaner? > > Mike > |
|
Oshel, Philip Eugene |
|
|
In reply to this post by Michael Giacomelli
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Mike, I've used both hydrogen peroxide and CLR (brand name lime & rust remover). For a TEM-sized chiller, this took one bottle of CLR or 1 Liter of 3% H2O2 (or 100 mL 30% H2O2). Circulate this for one to six hours, or even overnight if needed (sounds like that will be needed in your case), followed by flushing with 4-5 volumes of tap water. This works well, and causes no damage to instrument chilling lines. When I had a biofilm-sludge clogging problem, the vinegar trick worked. I used about 1 Liter of vinegar for this, followed by tap water. Phil On 11/23/2015 16:46 , Michael Giacomelli wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I recently discovered an old laser in storage. Unfortunately the > previous users had not drained the coolant lines, resulting in several > years corrosion blocking the lines. I flushed them out as best I > could and found a grey metal slurry that was fairly thick in place of > what was once the coolant. Unfortunately, even with flushing I can't > flow anything through the system using a normal thermotek chiller. > > Are there any options in this situation to unclog a laser? Looking at > the material in the line, its fairly diffuse and does not appear all > that solid. Is it possible to use something like a weak acid to > dissolve any solid material blocking the line? Or some other > solvent/cleaner? > > Mike > -- Philip Oshel Microscopy Facility Supervisor Biology Department 024C Brooks Hall Central Michigan University Mt. Pleasant, MI 48859 (989) 774-3576 |
|
Mark Cannell-2 |
|
|
In reply to this post by Craig Brideau
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I wonder if 2 part automotive radiator flush would be the best option? HTH Mark On 23/11/2015, at 10:23 pm, Craig Brideau <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I've used household vinegar for this, or you can make up 5% acetic acid > from lab stock. If you can get any flow at all, just keep running solution > through it. If you're lucky, the vinegar will dissolve out any mineral > deposits and also help loosen up any biological gunk. Once you have things > flowing well, flush with regular water. Finally, put Optishield or similar > into your final coolant water to help keep this from happening in the > future. I'm sure others will chime in with their own magic recipes for > loosening up stuff like this, so give it all a try and you should hit on > something that works. > > Best of luck! > > Craig Brideau > > On Mon, Nov 23, 2015 at 2:46 PM, Michael Giacomelli <[hidden email]> wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> I recently discovered an old laser in storage. Unfortunately the >> previous users had not drained the coolant lines, resulting in several >> years corrosion blocking the lines. I flushed them out as best I >> could and found a grey metal slurry that was fairly thick in place of >> what was once the coolant. Unfortunately, even with flushing I can't >> flow anything through the system using a normal thermotek chiller. >> >> Are there any options in this situation to unclog a laser? Looking at >> the material in the line, its fairly diffuse and does not appear all >> that solid. Is it possible to use something like a weak acid to >> dissolve any solid material blocking the line? Or some other >> solvent/cleaner? >> >> Mike >> Mark B. Cannell Ph.D. FRSNZ Professor of Cardiac Cell Biology School of Physiology & Pharmacology Faculty of Biomedical Sciences University of Bristol Bristol BS8 1TD UK [hidden email] |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** I'm not sure I would trust radiator flush. The metals and plastics used in the laser might be more delicate than those used in a car radiator. I would only resort to something like that if nothing else worked. Craig Brideau On Wed, Nov 25, 2015 at 5:46 PM, Mark Cannell <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I wonder if 2 part automotive radiator flush would be the best option? > > HTH > > Mark > > > On 23/11/2015, at 10:23 pm, Craig Brideau <[hidden email]> wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > I've used household vinegar for this, or you can make up 5% acetic acid > > from lab stock. If you can get any flow at all, just keep running > solution > > through it. If you're lucky, the vinegar will dissolve out any mineral > > deposits and also help loosen up any biological gunk. Once you have > things > > flowing well, flush with regular water. Finally, put Optishield or > similar > > into your final coolant water to help keep this from happening in the > > future. I'm sure others will chime in with their own magic recipes for > > loosening up stuff like this, so give it all a try and you should hit on > > something that works. > > > > Best of luck! > > > > Craig Brideau > > > > On Mon, Nov 23, 2015 at 2:46 PM, Michael Giacomelli <[hidden email]> > wrote: > > > >> ***** > >> To join, leave or search the confocal microscopy listserv, go to: > >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >> Post images on http://www.imgur.com and include the link in your > posting. > >> ***** > >> > >> I recently discovered an old laser in storage. Unfortunately the > >> previous users had not drained the coolant lines, resulting in several > >> years corrosion blocking the lines. I flushed them out as best I > >> could and found a grey metal slurry that was fairly thick in place of > >> what was once the coolant. Unfortunately, even with flushing I can't > >> flow anything through the system using a normal thermotek chiller. > >> > >> Are there any options in this situation to unclog a laser? Looking at > >> the material in the line, its fairly diffuse and does not appear all > >> that solid. Is it possible to use something like a weak acid to > >> dissolve any solid material blocking the line? Or some other > >> solvent/cleaner? > >> > >> Mike > >> > > Mark B. Cannell Ph.D. FRSNZ > Professor of Cardiac Cell Biology > School of Physiology & Pharmacology > Faculty of Biomedical Sciences > University of Bristol > Bristol > BS8 1TD UK > > [hidden email] > |
|
Mark Cannell-2 |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** Hi Craig Not sure about that, Al and Cu and plastics in engine too…. Cheers On 26/11/2015, at 1:09 am, Craig Brideau <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > I'm not sure I would trust radiator flush. The metals and plastics used in > the laser might be more delicate than those used in a car radiator. I would > only resort to something like that if nothing else worked. > > Craig Brideau > > On Wed, Nov 25, 2015 at 5:46 PM, Mark Cannell <[hidden email]> > wrote: > >> ***** >> To join, leave or search the confocal microscopy listserv, go to: >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >> Post images on http://www.imgur.com and include the link in your posting. >> ***** >> >> I wonder if 2 part automotive radiator flush would be the best option? >> >> HTH >> >> Mark >> >> >> On 23/11/2015, at 10:23 pm, Craig Brideau <[hidden email]> wrote: >> >>> ***** >>> To join, leave or search the confocal microscopy listserv, go to: >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>> Post images on http://www.imgur.com and include the link in your >> posting. >>> ***** >>> >>> I've used household vinegar for this, or you can make up 5% acetic acid >>> from lab stock. If you can get any flow at all, just keep running >> solution >>> through it. If you're lucky, the vinegar will dissolve out any mineral >>> deposits and also help loosen up any biological gunk. Once you have >> things >>> flowing well, flush with regular water. Finally, put Optishield or >> similar >>> into your final coolant water to help keep this from happening in the >>> future. I'm sure others will chime in with their own magic recipes for >>> loosening up stuff like this, so give it all a try and you should hit on >>> something that works. >>> >>> Best of luck! >>> >>> Craig Brideau >>> >>> On Mon, Nov 23, 2015 at 2:46 PM, Michael Giacomelli <[hidden email]> >> wrote: >>> >>>> ***** >>>> To join, leave or search the confocal microscopy listserv, go to: >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy >>>> Post images on http://www.imgur.com and include the link in your >> posting. >>>> ***** >>>> >>>> I recently discovered an old laser in storage. Unfortunately the >>>> previous users had not drained the coolant lines, resulting in several >>>> years corrosion blocking the lines. I flushed them out as best I >>>> could and found a grey metal slurry that was fairly thick in place of >>>> what was once the coolant. Unfortunately, even with flushing I can't >>>> flow anything through the system using a normal thermotek chiller. >>>> >>>> Are there any options in this situation to unclog a laser? Looking at >>>> the material in the line, its fairly diffuse and does not appear all >>>> that solid. Is it possible to use something like a weak acid to >>>> dissolve any solid material blocking the line? Or some other >>>> solvent/cleaner? >>>> >>>> Mike >>>> >> >> Mark B. Cannell Ph.D. FRSNZ >> Professor of Cardiac Cell Biology >> School of Physiology & Pharmacology >> Faculty of Biomedical Sciences >> University of Bristol >> Bristol >> BS8 1TD UK >> >> [hidden email] >> Mark B. Cannell Ph.D. FRSNZ Professor of Cardiac Cell Biology School of Physiology & Pharmacology Faculty of Biomedical Sciences University of Bristol Bristol BS8 1TD UK [hidden email] |
|
Craig Brideau |
|
|
*****
To join, leave or search the confocal microscopy listserv, go to: http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy Post images on http://www.imgur.com and include the link in your posting. ***** On a related note, I heard from my friendly neighbourhood technicians that Coherent is switching over to this stuff called CoolFlow IGE for the liquid cooling systems. It claims to have good corrosion resistance, although I haven't yet used it myself. I might give it a try during my next line flush and cleanout though. Craig On Thu, Nov 26, 2015 at 1:48 AM, Mark Cannell <[hidden email]> wrote: > ***** > To join, leave or search the confocal microscopy listserv, go to: > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > Post images on http://www.imgur.com and include the link in your posting. > ***** > > Hi Craig > > Not sure about that, Al and Cu and plastics in engine too…. > > Cheers > > On 26/11/2015, at 1:09 am, Craig Brideau <[hidden email]> wrote: > > > ***** > > To join, leave or search the confocal microscopy listserv, go to: > > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > > Post images on http://www.imgur.com and include the link in your > posting. > > ***** > > > > I'm not sure I would trust radiator flush. The metals and plastics used > in > > the laser might be more delicate than those used in a car radiator. I > would > > only resort to something like that if nothing else worked. > > > > Craig Brideau > > > > On Wed, Nov 25, 2015 at 5:46 PM, Mark Cannell < > [hidden email]> > > wrote: > > > >> ***** > >> To join, leave or search the confocal microscopy listserv, go to: > >> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >> Post images on http://www.imgur.com and include the link in your > posting. > >> ***** > >> > >> I wonder if 2 part automotive radiator flush would be the best option? > >> > >> HTH > >> > >> Mark > >> > >> > >> On 23/11/2015, at 10:23 pm, Craig Brideau <[hidden email]> > wrote: > >> > >>> ***** > >>> To join, leave or search the confocal microscopy listserv, go to: > >>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >>> Post images on http://www.imgur.com and include the link in your > >> posting. > >>> ***** > >>> > >>> I've used household vinegar for this, or you can make up 5% acetic acid > >>> from lab stock. If you can get any flow at all, just keep running > >> solution > >>> through it. If you're lucky, the vinegar will dissolve out any mineral > >>> deposits and also help loosen up any biological gunk. Once you have > >> things > >>> flowing well, flush with regular water. Finally, put Optishield or > >> similar > >>> into your final coolant water to help keep this from happening in the > >>> future. I'm sure others will chime in with their own magic recipes for > >>> loosening up stuff like this, so give it all a try and you should hit > on > >>> something that works. > >>> > >>> Best of luck! > >>> > >>> Craig Brideau > >>> > >>> On Mon, Nov 23, 2015 at 2:46 PM, Michael Giacomelli <[hidden email]> > >> wrote: > >>> > >>>> ***** > >>>> To join, leave or search the confocal microscopy listserv, go to: > >>>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy > >>>> Post images on http://www.imgur.com and include the link in your > >> posting. > >>>> ***** > >>>> > >>>> I recently discovered an old laser in storage. Unfortunately the > >>>> previous users had not drained the coolant lines, resulting in several > >>>> years corrosion blocking the lines. I flushed them out as best I > >>>> could and found a grey metal slurry that was fairly thick in place of > >>>> what was once the coolant. Unfortunately, even with flushing I can't > >>>> flow anything through the system using a normal thermotek chiller. > >>>> > >>>> Are there any options in this situation to unclog a laser? Looking at > >>>> the material in the line, its fairly diffuse and does not appear all > >>>> that solid. Is it possible to use something like a weak acid to > >>>> dissolve any solid material blocking the line? Or some other > >>>> solvent/cleaner? > >>>> > >>>> Mike > >>>> > >> > >> Mark B. Cannell Ph.D. FRSNZ > >> Professor of Cardiac Cell Biology > >> School of Physiology & Pharmacology > >> Faculty of Biomedical Sciences > >> University of Bristol > >> Bristol > >> BS8 1TD UK > >> > >> [hidden email] > >> > > Mark B. Cannell Ph.D. FRSNZ > Professor of Cardiac Cell Biology > School of Physiology & Pharmacology > Faculty of Biomedical Sciences > University of Bristol > Bristol > BS8 1TD UK > > [hidden email] > |
|
|
Coolflow IGE has been good for us — we’ve been using it in our Chameleon for about a year and a half, and it has reduced our clogging issues a lot. To switch from Optishield, you need to completely flush, rinse, and empty the lines, and install an inline mesh filter (which is the biggest up-front cost at $150; a big jug of the premixed solution is $54, over a year’s supply). Coherent sent us a step-by-step guide. Theresa On Dec 2, 2015, at 2:14 PM, Craig Brideau <[hidden email]<mailto:[hidden email]>> wrote: ***** On a related note, I heard from my friendly neighbourhood technicians that Coherent is switching over to this stuff called CoolFlow IGE for the liquid cooling systems. It claims to have good corrosion resistance, although I haven't yet used it myself. I might give it a try during my next line flush and cleanout though. Craig ------------------------------------ Theresa Swayne, Ph.D. Manager Confocal and Specialized Microscopy Shared Resource<http://hiccc.columbia.edu/research/sharedresources/confocal> Herbert Irving Comprehensive Cancer Center Columbia University Medical Center 1130 St. Nicholas Ave., Room 222A New York, NY 10032 Phone: 212-851-4613 [hidden email]<mailto:[hidden email]> |
|
Armstrong, Brian |
|
|
Hello, I used to have clogging problems and tried a few techniques: 1/3 white vinegar 2/3 distilled H2O (not MilliQ). I was told by my LASER Pro not to blow with too much force (he used oral pressure [blew with his mouth]). I got dizzy using this method so I used room air at light flow. I have special hoses that I cut/shortened that have the proper connector ends (male & female) for attachment to the LASER. I also used to change the distilled water on the first day of every month. This really helped.
However, once I switched to the CoolFlow and stopped using distilled water with Optishield I never had another problem with flow, temperature or clogging. I highly recommend switching. It costs a little but saves a lot of time and aggravation. Cheers, Brian D Armstrong PhD Associate Research Professor Director, Light Microscopy Core Beckman Research Institute City of Hope 1450 E Duarte Rd Duarte, CA 91010 626-256-4673 x62872 -----Original Message----- From: Confocal Microscopy List [mailto:[hidden email]] On Behalf Of Swayne, Theresa C. Sent: Wednesday, December 02, 2015 11:56 AM To: [hidden email] Subject: Re: Unclogging corroded laser coolant lines Coolflow IGE has been good for us — we’ve been using it in our Chameleon for about a year and a half, and it has reduced our clogging issues a lot. To switch from Optishield, you need to completely flush, rinse, and empty the lines, and install an inline mesh filter (which is the biggest up-front cost at $150; a big jug of the premixed solution is $54, over a year’s supply). Coherent sent us a step-by-step guide. Theresa On Dec 2, 2015, at 2:14 PM, Craig Brideau <[hidden email]<mailto:[hidden email]>> wrote: ***** On a related note, I heard from my friendly neighbourhood technicians that Coherent is switching over to this stuff called CoolFlow IGE for the liquid cooling systems. It claims to have good corrosion resistance, although I haven't yet used it myself. I might give it a try during my next line flush and cleanout though. Craig ------------------------------------ Theresa Swayne, Ph.D. Manager Confocal and Specialized Microscopy Shared Resource<http://hiccc.columbia.edu/research/sharedresources/confocal> Herbert Irving Comprehensive Cancer Center Columbia University Medical Center 1130 St. Nicholas Ave., Room 222A New York, NY 10032 Phone: 212-851-4613 [hidden email]<mailto:[hidden email]> --------------------------------------------------------------------- *SECURITY/CONFIDENTIALITY WARNING: This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. (fpc5p) --------------------------------------------------------------------- |